Re: widefield autofluorescence in unlabeled cells - a filter mystery?

Posted by Chen, De (NIH/NCI) [C]-2 on
URL: http://confocal-microscopy-list.275.s1.nabble.com/widefield-autofluorescence-in-unlabeled-cells-a-filter-mystery-tp7592150p7592154.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

It looks like it is the auto fluorescence from the cell. Do you see Red (561nm excitation) or far red (647nm excitation) fluorescence? This could be due to the cell culture issue often encountered in a biology lab, do you feel difficulty in transfecting cells?  Best try new cell line, new media and make sure the incubator is running under normal condition, etc.




-----Original Message-----
From: Cromey, Douglas W - (dcromey) <[hidden email]>
Sent: Monday, April 26, 2021 12:48 PM
To: [hidden email]
Subject: widefield autofluorescence in unlabeled cells - a filter mystery?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

A colleague asked me to take a look at their widefield microscope. It is an inverted microscope with a 100W Hg source, excitation filter wheel, a couple of choices for dichroics in the microscope filter changer and a filter wheel in front of the camera. They are seeing unlabeled cells fluoresce green (FITC/GFP set) with an otherwise black background where there are no cells. The microscope is approximately 15 years old.

My guess is that the excitation filters have failed (or are failing) after being on the receiving end of a 100W Hg lamp for all this time. Any other thoughts?

Doug

^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^