Hi Doug,
picture = 1000 words. Post a (link to a) picture.
It could be autofluorescence. Do the unlabeled cells look green on other
widefield fluorescence microscopes?
It could be transmission. Block all light between the sample and the
condenser. Does that help? Even a room light on the ceiling could work as
darkfield illumination. Btw, what's the objective lens?
It does not look like filter bleed through. That usually looks like (quite
non-uniform) signal independent of the cells.
Are the cells live or fixed? In PBS or in index-matching mounting medium?
Best, zdenek
On Mon, Apr 26, 2021 at 12:50 PM Cromey, Douglas W - (dcromey) <
[hidden email]> wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> A colleague asked me to take a look at their widefield microscope. It is
> an inverted microscope with a 100W Hg source, excitation filter wheel, a
> couple of choices for dichroics in the microscope filter changer and a
> filter wheel in front of the camera. They are seeing unlabeled cells
> fluoresce green (FITC/GFP set) with an otherwise black background where
> there are no cells. The microscope is approximately 15 years old.
>
> My guess is that the excitation filters have failed (or are failing) after
> being on the receiving end of a 100W Hg lamp for all this time. Any other
> thoughts?
>
> Doug
>
> ^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^
> Douglas W. Cromey, M.S. - Associate Scientific Investigator
> Dept. of Cellular & Molecular Medicine, University of Arizona
> Life Sciences North, Room 463
> 1333 N. Martin Ave, Tucson, AZ 85721 USA
>
> office: LSN 463 email:
[hidden email]<mailto:
>
[hidden email]>
> voice: 520-626-2824 fax: 520-626-2097
>
> UA Microscopy Alliance -
http://microscopy.arizona.edu/> A collaborative effort to bring information about shared microscopy
> facilities to the University of Arizona and the community.
>
--
--
Zdenek Svindrych, Ph.D.
Research Scientist - Microscopy Imaging Specialist
Department of Biochemistry and Cell Biology
Geisel School of Medicine at Dartmouth