> There are submicrosecond level shifters such as this:
>
https://www.onsemi.com/pdf/datasheet/mc14504b-d.pdf, and if the impedance
> was high enough on the input you could even use a high side driver for even
> faster switching speeds. However, if you are driving a Pockels cell (which
> normally requires analog instead of digital control) I would just use an
> op-amp voltage multiplier. A unity gain stable op-amp would be the easiest
> bet, and one like this would be plenty fast:
>
>
https://www.digikey.com/en/products/detail/analog-devices-inc/LT6200CS6-TRMPBF/1115933>
> On Sat, May 8, 2021 at 12:58 PM Khoroshyy Petro <
[hidden email]>
> wrote:
>
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> >
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> > Post images on
http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > Hi everybody.
> >
> > This question is not directly about confocal microscopy but, and might be
> > better addressed on some electrical engineering forum. Bit i believe
> there
> > is someone who can guide me in the right direction.
> > I need to drive a Pockel cell driver with 10 V pulses, but available
> pulse
> > generator has max 5V output. I need pulses with submicrosecond rising
> time.
> > I have searched for level shifters chips with needed bandwidth and
> > impedance but all of what I found work up to 5.5 volts.
> > I wonder if someone can recommend some solution for the problem or give a
> > hint.
> > Thanks.
> > Petro.
> >
>
>
> --