For these tasks an SRS SR560 comes in very handy, which can also be tailored for bandwidth. (No commercial interest, just a really useful multi-use lab accessory). But there are other excellent choices, too.
> Am 09.05.2021 um 14:24 schrieb Mark Cannell <
[hidden email]>:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Maybe but it apparently needs a slew rate > 10V/us into what may be a 50 ohm impedance. This is RF circuitry. Perhaps the manufacturer can give some more details as it's hard to define what's needed on the basis of the supplied info. IMHO
>
> Cheers
>
> Mark B. Cannell. Ph.D. FRSNZ FISHR
> Department of Physiology, Pharmacology & Neuroscience
> School of Medical Sciences
> University Walk
> Bristol BS8 1TD
>
>
[hidden email]
>
>
>
> On 9/05/21, 10:36 AM, "Confocal Microscopy List on behalf of Nino Karpf" <
[hidden email] on behalf of
[hidden email]> wrote:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> Post images on
http://www.imgur.com and include the link in your posting.
> *****
>
> Dear Petro,
>
> Would an amplifier solve your problem? If you buy an electronic amplifier that amplifies your driving voltage 2-3 times (from 5V to 10V or higher) you should be able to drive your pockels cell. Unless I’m thinking too easy here. You can buy suitable amplifiers from e.g. Minicircuits.
>
> Best,
> Nino
>
> Von meinem iPhone gesendet
>
>> Am 09.05.2021 um 10:44 schrieb Khoroshyy Petro <
[hidden email]>:
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>> Post images on
http://www.imgur.com and include the link in your posting.
>> *****
>>
>> Thanks.
>> I will have a look on the shifter.
>> I am a bit lost by analog control?
>> Do you mean setting up the HT by control voltage?
>> Thanks.
>> Petro.
>>
>>> On Sat, May 8, 2021, 23:12 Benjamin Smith <
[hidden email]>
>>> wrote:
>>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>>> Post images on
http://www.imgur.com and include the link in your posting.
>>> *****
>>>
>>> There are submicrosecond level shifters such as this:
>>>
https://www.onsemi.com/pdf/datasheet/mc14504b-d.pdf, and if the impedance
>>> was high enough on the input you could even use a high side driver for even
>>> faster switching speeds. However, if you are driving a Pockels cell (which
>>> normally requires analog instead of digital control) I would just use an
>>> op-amp voltage multiplier. A unity gain stable op-amp would be the easiest
>>> bet, and one like this would be plenty fast:
>>>
>>>
https://www.digikey.com/en/products/detail/analog-devices-inc/LT6200CS6-TRMPBF/1115933>>>
>>> On Sat, May 8, 2021 at 12:58 PM Khoroshyy Petro <
[hidden email]>
>>> wrote:
>>>
>>>> *****
>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>>>> Post images on
http://www.imgur.com and include the link in your
>>> posting.
>>>> *****
>>>>
>>>> Hi everybody.
>>>>
>>>> This question is not directly about confocal microscopy but, and might be
>>>> better addressed on some electrical engineering forum. Bit i believe
>>> there
>>>> is someone who can guide me in the right direction.
>>>> I need to drive a Pockel cell driver with 10 V pulses, but available
>>> pulse
>>>> generator has max 5V output. I need pulses with submicrosecond rising
>>> time.
>>>> I have searched for level shifters chips with needed bandwidth and
>>>> impedance but all of what I found work up to 5.5 volts.
>>>> I wonder if someone can recommend some solution for the problem or give a
>>>> hint.
>>>> Thanks.
>>>> Petro.
>>>>
>>>
>>>
>>> --
>>> Benjamin E. Smith, Ph. D.
>>> Imaging Specialist, Vision Science
>>> University of California, Berkeley
>>> 195 Weill Hall
>>> Berkeley, CA 94720-3200
>>> Tel (510) 642-9712
>>> Fax (510) 643-6791
>>> e-mail:
[hidden email]
>>>
>>>
https://vision.berkeley.edu/faculty/core-grants-nei/core-grant-microscopic-imaging/>>>
>