Re: 5 to 10V level shifter?
Posted by
Jan-Peter Siebrasse on
URL: http://confocal-microscopy-list.275.s1.nabble.com/5-to-10V-level-shifter-tp7592213p7592223.html
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopyPost images on
http://www.imgur.com and include the link in your posting.
*****
Hi Pedro,
I had a similar problem with an old AOTF which needed 10V but my homebrew controller based on an Arduino microcontroller would only deliver 5V max.
As a workaround I used a simple circuit with the TL084CN op-amp from Texas instrument, combined with the d/a converter MCP4822. So the Arduino now deliver 0-5V via the MCP4822 which is translated to 0-10V by the op-amp.
Time resolution should be around 10µs but I never tested it.
Regards
JP