Hi Alison,
It's funny you should say that because I've just learned it here:
https://www.microscope.healthcare.nikon.com/en_EU/learn/video-tutorialsChoose "Objective lenses and Optical Aberrations" and go to 15.04.
According to Nikon, it's not just refractive index that matters, but also dispersion that must be matched to the lens glass...
Sincerely,
Matthieu
-----Original Message-----
From: Confocal Microscopy List [mailto:
[hidden email]] On Behalf Of Alison J. North
Sent: 21 May 2021 19:04
To:
[hidden email]
Subject: Re: Experience with Olympus X-Line 100X objective?
This email was sent to you by someone outside the University.
You should only click on links or attachments if you are certain that the email is genuine and the content is safe.
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopyPost images on
http://www.imgur.com and include the link in your posting.
*****
Hi Abby,
So the situation with the Olympus objectives and different oils has been an interesting journey for me. And I'm happy to have an opportunity to publicly set the records straight because I previously commented that the Olympus 100x SApo - which is described as being corrected across all regular wavelength dyes - appeared to give chromatic shifts. Like you, I use DeltaVisions and an OMX a lot. (Of course, you don't really notice the chromatic aberrations with the OMX because you use the image registration step to correct for the multiple cameras anyway.) With the DeltaVisions we always had to correct for chromatic shift when imaging at high resolution, e.g. for yeast or bacteria, and I blamed this entirely on the objectives. It never occurred to me that the Cargille oils, which we always use to compensate for refractive index mismatch, could be the issue, until somebody at GE told me a couple of years ago that it seemed that the composition of the Cargille oils had changed over the past few years and now they appeared to induce chromatic dispersion. Therefore, when I tested the new X-line 60x objective against the old 60x, I did so both using Cargille oil and standard Olympus oil, and that's how I confirmed that indeed the shift was far more pronounced with the Cargille oil for the old objective.
Of course, this leaves one with a bit of a dilemma - I would still want to use the Cargille oils because of their fabulous range in r.i. that allows you to minimize spherical aberrations. So I was delighted to find that with the X-line 60x I can stick to the Cargille oils without worrying about chromatic aberrations. And I owe a public apology to Olympus for having maligned their SApo objectives unfairly!
I do think it's an important lesson. People are all too happy to switch oils around without having thoroughly tested the consequences.
I'd be interested to hear how you get on with the 100x/1.45 X-line objective - I'm certainly sold on the 40x and 60x versions!
All the best,
Alison
________________________________
From: Confocal Microscopy List <
[hidden email]> on behalf of Abby Dernburg PhD <
[hidden email]>
Sent: Friday, May 21, 2021 11:02 AM
To:
[hidden email] <
[hidden email]>
Subject: Re: Experience with Olympus X-Line 100X objective?
*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFaQ&c=JeTkUgVztGMmhKYjxsy2rfoWYibK1YmxXez1G3oNStg&r=RBx0-WJrAO5vwSOLNmFbqYvikvIZS5ns3-USwvMOuLo&m=HW0j46tfOJaTyP8RjuJ5ZP_MfnqTN-q-huPyoNULrE0&s=QNZ1Vxc7NHyZ7J-pr29txEQeUdZo1Q8N09PPFl4KVEg&e=Post images on
https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFaQ&c=JeTkUgVztGMmhKYjxsy2rfoWYibK1YmxXez1G3oNStg&r=RBx0-WJrAO5vwSOLNmFbqYvikvIZS5ns3-USwvMOuLo&m=HW0j46tfOJaTyP8RjuJ5ZP_MfnqTN-q-huPyoNULrE0&s=oViXu4_HYXZj9AXrvYXRTNWfO1-HND9lhMlpp03OmzE&e= and include the link in your posting.
*****
Thanks to those who responded - now I’m excited to try out the X-line (and thinking I should consider the 60X as well). I didn’t realize that Olympus oil might have been a better choice than the Cargille oils we’ve been using - thanks for that tip, Alison. We do see pretty severe chromatic dispersion in Z with our current 100X objective.
Abby Dernburg, Ph.D.
HHMI Investigator
Department of Molecular and Cell Biology University of California, Berkeley Berkeley, CA 94720 USA
Mobile: +1 510.427.0199
> On May 21, 2021, at 3:30 AM, Owens, Peter <
[hidden email]> wrote:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-> 2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFaQ&c=JeTkUgVztGMmhKYjxsy2rfoW
> YibK1YmxXez1G3oNStg&r=RBx0-WJrAO5vwSOLNmFbqYvikvIZS5ns3-USwvMOuLo&m=HW
> 0j46tfOJaTyP8RjuJ5ZP_MfnqTN-q-huPyoNULrE0&s=QNZ1Vxc7NHyZ7J-pr29txEQeUd
> Zo1Q8N09PPFl4KVEg&e= Post images on
>
https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFaQ&c=JeTkUgVztGMmhKYjxsy2rfoWYibK1YmxXez1G3oNStg&r=RBx0-WJrAO5vwSOLNmFbqYvikvIZS5ns3-USwvMOuLo&m=HW0j46tfOJaTyP8RjuJ5ZP_MfnqTN-q-huPyoNULrE0&s=oViXu4_HYXZj9AXrvYXRTNWfO1-HND9lhMlpp03OmzE&e= and include the link in your posting.
> *****
>
> I have just purchased a 100x xline for our facility in NUIG , to be delivered in a few weeks . Interesting to read your comments Alison on the psf and chromatic aberration. Certainly looking forward to trying it out and I can report back to this forum when i have a chance to use it. I have also heard that Olympus pricing for optics is due to increase by around 20% in June , and this was already a pricey objective.
>
> Best regards
>
> Peter
>
> www.imaging.nuigalway.ie<
https://urldefense.proofpoint.com/v2/url?u=ht> tp-3A__www.imaging.nuigalway.ie&d=DwIFaQ&c=JeTkUgVztGMmhKYjxsy2rfoWYib
> K1YmxXez1G3oNStg&r=RBx0-WJrAO5vwSOLNmFbqYvikvIZS5ns3-USwvMOuLo&m=HW0j46tfOJaTyP8RjuJ5ZP_MfnqTN-q-huPyoNULrE0&s=98RawYRKoHtC_g9VjPzmyX23HHHO_Ppk2sf1aEY0dNE&e=<
http://www.imaging.nuigalway.ie<
https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imaging.nuigalway.ie&d=DwIFaQ&c=JeTkUgVztGMmhKYjxsy2rfoWYibK1YmxXez1G3oNStg&r=RBx0-WJrAO5vwSOLNmFbqYvikvIZS5ns3-USwvMOuLo&m=HW0j46tfOJaTyP8RjuJ5ZP_MfnqTN-q-huPyoNULrE0&s=98RawYRKoHtC_g9VjPzmyX23HHHO_Ppk2sf1aEY0dNE&e=> > Centre for Microscopy and Imaging at NUI Galway - current research<
https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imaging.nuigalway.ie_&d=DwIFaQ&c=JeTkUgVztGMmhKYjxsy2rfoWYibK1YmxXez1G3oNStg&r=RBx0-WJrAO5vwSOLNmFbqYvikvIZS5ns3-USwvMOuLo&m=HW0j46tfOJaTyP8RjuJ5ZP_MfnqTN-q-huPyoNULrE0&s=2eyZOxnAR68xtkfRupUZaUnuhMXgQ7RLJYoQ8xGvpxQ&e= > The Centre for Microscopy and Imaging at NUIG (CMI) is a core facility dedicated to providing scientists with a variety of high-end light and electron microscopes along cutting-edge imaging software and analysis tools. Provide expertise in microscopy, imaging and analysis specialising in biomaterials, cell and tissue phenotypes.
> www.imaging.nuigalway.ie<
http://www.imaging.nuigalway.ie>
>
>
>
>
> ________________________________
> From: Confocal Microscopy List <
[hidden email]> on
> behalf of Alison J. North <
[hidden email]>
> Sent: 21 May 2021 11:03
> To:
[hidden email]
> <
[hidden email]>
> Subject: Re: Experience with Olympus X-Line 100X objective?
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-> 2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFaQ&c=JeTkUgVztGMmhKYjxsy2rfoW
> YibK1YmxXez1G3oNStg&r=RBx0-WJrAO5vwSOLNmFbqYvikvIZS5ns3-USwvMOuLo&m=HW
> 0j46tfOJaTyP8RjuJ5ZP_MfnqTN-q-huPyoNULrE0&s=QNZ1Vxc7NHyZ7J-pr29txEQeUd
> Zo1Q8N09PPFl4KVEg&e= Post images on
>
https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFaQ&c=JeTkUgVztGMmhKYjxsy2rfoWYibK1YmxXez1G3oNStg&r=RBx0-WJrAO5vwSOLNmFbqYvikvIZS5ns3-USwvMOuLo&m=HW0j46tfOJaTyP8RjuJ5ZP_MfnqTN-q-huPyoNULrE0&s=oViXu4_HYXZj9AXrvYXRTNWfO1-HND9lhMlpp03OmzE&e= and include the link in your posting.
> *****
>
> Hi Abby,
>
> I don't know whether this will be useful information or not, but we recently purchased the 60X X-line objective for our DeltaVision, and I can confirm that the psfs look great and the chromatic aberrations are significantly better than with the original 60x/1.42 oil objective, even when using the Cargille oils (which result in noticeable chromatic dispersion in comparison to when using Olympus oil). If the 100x X-line objective is anything like as good, then I really should be thinking about buying one myself - with the new 60x we no longer need to do any post-acquisition corrections.
>
> All the best,
> Alison
>
> Alison J. North, Ph.D.
> Senior Director of the Bio-Imaging Resource Center and Research
> Associate Professor The Rockefeller University
> 1230 York Avenue
> New York
> NY 10065
> Tel. 212 327 7488 (office - direct line)
>
https://www.rockefeller.edu/bioimaging/> RRID:SCR_017791
>
> ________________________________
> From: Confocal Microscopy List <
[hidden email]> on
> behalf of Abby Dernburg <
[hidden email]>
> Sent: Thursday, May 20, 2021 3:20 PM
> To:
[hidden email]
> <
[hidden email]>
> Subject: Experience with Olympus X-Line 100X objective?
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-> 2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFaQ&c=JeTkUgVztGMmhKYjxsy2rfoW
> YibK1YmxXez1G3oNStg&r=RBx0-WJrAO5vwSOLNmFbqYvikvIZS5ns3-USwvMOuLo&m=Nm
> NoIJspXtyELNquCYzD3rsekE-Q2HTeMGZalX8tkAE&s=PXZ7op5XSrDNBV28-AjO-FAY-P
> MyeGqCS8w3UOpmlAQ&e= Post images on
>
https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFaQ&c=JeTkUgVztGMmhKYjxsy2rfoWYibK1YmxXez1G3oNStg&r=RBx0-WJrAO5vwSOLNmFbqYvikvIZS5ns3-USwvMOuLo&m=NmNoIJspXtyELNquCYzD3rsekE-Q2HTeMGZalX8tkAE&s=8bFDOVYQ_MjyauSQQ8n-kZzEw04Py7NnwYipjGLnjM0&e= and include the link in your posting.
> *****
>
> Hi all -
>
> I am thinking of adding an Olympus X-Line 100X NA 1.45 to our DeltaVision Elite system and am curious about real-world experience with this objective, particularly for wide-field deconvolution microscopy. I haven't found any descirptions of performance other than marketing materials so I'm wondering how it measures up. Thanks in advance for any feedback!
>
> -Abby
The University of Edinburgh is a charitable body, registered in Scotland, with registration number SC005336. Is e buidheann carthannais a th’ ann an Oilthigh Dhùn Èideann, clàraichte an Alba, àireamh clàraidh SC005336.