Dear all,

I have an issue with ZDC unit (the old version ZDC, i.e. ZDC 1.) in our
IX81 TIRF microscope. I used it with MicroManager and it seems working.

I could manually find the focus and then find the offset, it looks working.
However when I turn on the autofocus, it is moving focus and defocusing
about 3um.
I tried to manually change the offset, but it does not take any effect.
The autofocus will focus ~3um away from the real focus.

Any suggestions? Thank you very much.


--
Best Regards,
Hongtao Chen
Straatman, Kees (Dr.) Straatman, Kees (Dr.)
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: uManager with Olympus IX81-ZDC

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear Hongtao Chen,

Sounds like you did not implemented the offset and it finds the bottom of your sample. Did you have a look at: https://micro-manager.org/wiki/Olympus, last few lines.

Best wishes

Kees


Dr Ir K.R. Straatman
Senior Experimental Officer
Advanced Imaging Facility
Centre for Core Biotechnology Services
University of Leicester
www.le.ac.uk/advanced-imaging-facility


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Hongtao Chen
Sent: 05 April 2018 19:07
To: [hidden email]
Subject: uManager with Olympus IX81-ZDC

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear all,

I have an issue with ZDC unit (the old version ZDC, i.e. ZDC 1.) in our
IX81 TIRF microscope. I used it with MicroManager and it seems working.

I could manually find the focus and then find the offset, it looks working.
However when I turn on the autofocus, it is moving focus and defocusing about 3um.
I tried to manually change the offset, but it does not take any effect.
The autofocus will focus ~3um away from the real focus.

Any suggestions? Thank you very much.


--
Best Regards,
Hongtao Chen
Hongtao Chen Hongtao Chen
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: uManager with Olympus IX81-ZDC

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi, Kees,

Thanks for reply.

Well, I did actually. I do have the searching range 500 and use
"measureoffset now"  to find the offset, it usually gives 2.xx
Then when I turned on the autofocus. It will defocus and the zdc actually
find focus 3um above the real focus plane I need.
I then tried to manually play the offset (plus, minus, larger, smaller
numbers),
it does not take any effect on "autofocus on", still the same 3um above.

Thanks.
Best Regards
Hongtao Chen

On Fri, Apr 6, 2018 at 12:40 AM, Straatman, Kees (Dr.) <[hidden email]
> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Dear Hongtao Chen,
>
> Sounds like you did not implemented the offset and it finds the bottom of
> your sample. Did you have a look at: https://micro-manager.org/
> wiki/Olympus, last few lines.
>
> Best wishes
>
> Kees
>
>
> Dr Ir K.R. Straatman
> Senior Experimental Officer
> Advanced Imaging Facility
> Centre for Core Biotechnology Services
> University of Leicester
> www.le.ac.uk/advanced-imaging-facility
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]]
> On Behalf Of Hongtao Chen
> Sent: 05 April 2018 19:07
> To: [hidden email]
> Subject: uManager with Olympus IX81-ZDC
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Dear all,
>
> I have an issue with ZDC unit (the old version ZDC, i.e. ZDC 1.) in our
> IX81 TIRF microscope. I used it with MicroManager and it seems working.
>
> I could manually find the focus and then find the offset, it looks working.
> However when I turn on the autofocus, it is moving focus and defocusing
> about 3um.
> I tried to manually change the offset, but it does not take any effect.
> The autofocus will focus ~3um away from the real focus.
>
> Any suggestions? Thank you very much.
>
>
> --
> Best Regards,
> Hongtao Chen
>



--
Best Regards,
Hongtao Chen