Hiroyuki Hakozaki |
|
|
Dear Confocal list members,
This isn't really confocal related question but hope some people on this list has experience with DPSS 561nm laser. I'm considering to purchase DPSS 561nm laser - 100mW or 150mW for my microscope application. I looked at Coherent, Spectra Physics, CrystaLaser and Cobolt. Since this kind of laser is new to the market, I would like to hear real experience with this laser. If you have experience with this laser, can you share your opinion? Best Regards, Hiro. ======================================= Hiroyuki Hakozaki National Center for Microscopy & Imaging Research University of California San Diego ======================================= |
|
Craig Brideau |
|
|
561 has been around for a while. Just make sure the output mode is
good and the laser is power stabilized (check RMS noise on the output). Are you planing on fiber coupling it or steering the beam with mirrors (freespace)? Craig On Tue, Oct 13, 2009 at 4:28 PM, Hiroyuki Hakozaki <[hidden email]> wrote: > Dear Confocal list members, > This isn't really confocal related question but hope some people on > this list has experience with DPSS 561nm laser. > > I'm considering to purchase DPSS 561nm laser - 100mW or 150mW for my > microscope application. I looked at Coherent, Spectra Physics, > CrystaLaser and Cobolt. > > Since this kind of laser is new to the market, I would like to hear > real experience with this laser. > > If you have experience with this laser, can you share your opinion? > > Best Regards, > > Hiro. > ======================================= > Hiroyuki Hakozaki > > National Center for Microscopy & Imaging Research > University of California San Diego > ======================================= > |
|
Hiroyuki Hakozaki |
|
|
Hi Craig,
Thank you for your information. As for specification wise, all of them looks fine. I'm planing to use fiber coupling. Hiro. > 561 has been around for a while. Just make sure the output mode is > good and the laser is power stabilized (check RMS noise on the > output). Are you planing on fiber coupling it or steering the beam > with mirrors (freespace)? > > Craig > > > On Tue, Oct 13, 2009 at 4:28 PM, Hiroyuki Hakozaki > <[hidden email]> wrote: > > Dear Confocal list members, > > ?This isn't really confocal related question but hope some people on > > this list has experience with DPSS 561nm laser. > > > > ?I'm considering to purchase DPSS 561nm laser - 100mW or 150mW for my > > microscope application. I looked at Coherent, Spectra Physics, > > CrystaLaser and Cobolt. > > > > ?Since this kind of laser is new to the market, I would like to hear > > real experience with this laser. > > > > ?If you have experience with this laser, can you share your opinion? > > > > Best Regards, > > > > Hiro. > > ======================================= > > Hiroyuki Hakozaki > > > > National Center for Microscopy & Imaging Research > > University of California San Diego > > ======================================= > > |
|
Alberto Diaspro |
|
|
In reply to this post by Hiroyuki Hakozaki
Friends
I am intersted as well Ciao ALby On Oct 14, 2009, at 12:28 AM, Hiroyuki Hakozaki wrote: > Dear Confocal list members, > This isn't really confocal related question but hope some people on > this list has experience with DPSS 561nm laser. > > I'm considering to purchase DPSS 561nm laser - 100mW or 150mW for my > microscope application. I looked at Coherent, Spectra Physics, > CrystaLaser and Cobolt. > > Since this kind of laser is new to the market, I would like to hear > real experience with this laser. > > If you have experience with this laser, can you share your opinion? > > Best Regards, > > Hiro. > ======================================= > Hiroyuki Hakozaki > > National Center for Microscopy & Imaging Research > University of California San Diego > ======================================= ---------------------------------------- Alberto Diaspro Head, Nanophysics Unit Senior Scientist The Italian Institute of Technology -IIT Via Morego, 30 16163 - Genova (Italy) phone: +39 010 71781503 mobile: +393666719968 fax: +39 010 720321 http://www.iit.it [hidden email] Professor of Applied Physics Department of Physics University of Genova Via Dodecaneso, 33 16146 Genova - Italy tel. +39 010 353 6426 fax. +39 010 314218 http://www.lambs.it [hidden email] ------------------------------------------------------- |
|
Alberto Diaspro |
|
|
In reply to this post by Craig Brideau
Data sheet look good for the Coherent one.
On Oct 14, 2009, at 2:00 AM, Craig Brideau wrote: > 561 has been around for a while. Just make sure the output mode is > good and the laser is power stabilized (check RMS noise on the > output). Are you planing on fiber coupling it or steering the beam > with mirrors (freespace)? > > Craig > > > On Tue, Oct 13, 2009 at 4:28 PM, Hiroyuki Hakozaki > <[hidden email]> wrote: >> Dear Confocal list members, >> This isn't really confocal related question but hope some people on >> this list has experience with DPSS 561nm laser. >> >> I'm considering to purchase DPSS 561nm laser - 100mW or 150mW for my >> microscope application. I looked at Coherent, Spectra Physics, >> CrystaLaser and Cobolt. >> >> Since this kind of laser is new to the market, I would like to hear >> real experience with this laser. >> >> If you have experience with this laser, can you share your opinion? >> >> Best Regards, >> >> Hiro. >> ======================================= >> Hiroyuki Hakozaki >> >> National Center for Microscopy & Imaging Research >> University of California San Diego >> ======================================= >> ---------------------------------------- Alberto Diaspro Head, Nanophysics Unit Senior Scientist The Italian Institute of Technology -IIT Via Morego, 30 16163 - Genova (Italy) phone: +39 010 71781503 mobile: +393666719968 fax: +39 010 720321 http://www.iit.it [hidden email] Professor of Applied Physics Department of Physics University of Genova Via Dodecaneso, 33 16146 Genova - Italy tel. +39 010 353 6426 fax. +39 010 314218 http://www.lambs.it [hidden email] ------------------------------------------------------- |
|
|
In reply to this post by Hiroyuki Hakozaki
We have a 100mW laser from Cobolt for approximately 1 year now. We use
it on a spinning disk system and it works fine (although its power was never really 100 mW, rather 90-95). Cheers Gabor |
|
Craig Brideau |
|
|
In reply to this post by Hiroyuki Hakozaki
Single-mode fiber coupling tends to clean up the mode output of your
laser as long as it is done correctly (it is a form of spatial filtering). You should have a clean gaussian mode out of the single-mode fiber tip. The next parameter is the power stability of the laser. You need to know the RMS noise in the power output of the laser, particularly the noise at frequencies similar to your image/pixel acquisition rate. A 'decent' laser has RMS noise <5% and will be usable for most applications. A 'good' laser is <1-2%. Unless you are trying photobleaching or the like maximum power is seldom an issue; for most people 100mW is overkill. Craig 2009/10/13 Hiroyuki Hakozaki <[hidden email]>: > Hi Craig, > Thank you for your information. As for specification wise, all of them > looks fine. > > I'm planing to use fiber coupling. > > Hiro. > >> 561 has been around for a while. Just make sure the output mode is >> good and the laser is power stabilized (check RMS noise on the >> output). Are you planing on fiber coupling it or steering the beam >> with mirrors (freespace)? >> >> Craig >> >> >> On Tue, Oct 13, 2009 at 4:28 PM, Hiroyuki Hakozaki >> <[hidden email]> wrote: >> > Dear Confocal list members, >> > ?This isn't really confocal related question but hope some people on >> > this list has experience with DPSS 561nm laser. >> > >> > ?I'm considering to purchase DPSS 561nm laser - 100mW or 150mW for my >> > microscope application. I looked at Coherent, Spectra Physics, >> > CrystaLaser and Cobolt. >> > >> > ?Since this kind of laser is new to the market, I would like to hear >> > real experience with this laser. >> > >> > ?If you have experience with this laser, can you share your opinion? >> > >> > Best Regards, >> > >> > Hiro. >> > ======================================= >> > Hiroyuki Hakozaki >> > >> > National Center for Microscopy & Imaging Research >> > University of California San Diego >> > ======================================= >> > > |
|
Hiroyuki Hakozaki |
|
|
In reply to this post by Hiroyuki Hakozaki
Dear All,
Thank you for your input about DPSS 561nm laser. I also received private responses from my friends and would like to summarize here. One of my friend have very good experience include support with Coherent 100mW which he has more than one. I also heard good and bad experience with Cobolt. I didn't hear anything about CrystaLaser and Spectra Physics. So no comment about these company. Because of my friend have good experience with Coherent, I will go with Coherent. Thank you again for your feedback. Best Regards, Hiro. > Dear Confocal list members, > This isn't really confocal related question but hope some people on > this list has experience with DPSS 561nm laser. > > I'm considering to purchase DPSS 561nm laser - 100mW or 150mW for my > microscope application. I looked at Coherent, Spectra Physics, > CrystaLaser and Cobolt. > > Since this kind of laser is new to the market, I would like to hear > real experience with this laser. > > If you have experience with this laser, can you share your opinion? > > Best Regards, > > Hiro. > ======================================= > Hiroyuki Hakozaki > > National Center for Microscopy & Imaging Research > University of California San Diego > ======================================= |