Hi all,
I'm looking for some trustworthy, permanent standard(s) to use in a MP system to monitor the condition of detectors and the system in general. The goal is a fluorescent source that is immutable. This could done either by having unrestricted replacement of bleached label with unbleached label (i.e. a solution), or a uniform and very large solid, so that it is unlikely that any one tiny volume will be imaged repeatedly. We've tried several iterations of an aqueous solution of label, but sealing the prep does not prevent eventual evaporation. Plus, there seems to be some modification of FITC, for example, in solution over time (a month) regardless of exposure to light. I'm not sure I can trust the consistency of colored plastic slides, since they are not manufactured with this role in mind. One possibility is uranyl glass slides, but Corning says they don't make these any more. Does anyone know a source? We're also considering quantum dots in something like immersion oil , in a sealed chamber but I don't have any experience with these labels in a hydrophobic environment. Thanks for your insights. C Carl A. Boswell, Ph.D. Molecular and Cellular Biology Univ. of Arizona 520-954-7053 FAX 520-621-3709 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** How about Fluorescent plastic slides -I got some free from Chroma. Cheers On 15/01/2011, at 12:34 PM, Boswell, Carl A - (cboswell) wrote: > Hi all, > I'm looking for some trustworthy, permanent standard(s) to use in a > MP system to monitor the condition of detectors and the system in > general. The goal is a fluorescent source that is immutable. This > could done either by having unrestricted replacement of bleached > label with unbleached label (i.e. a solution), or a uniform and very > large solid, so that it is unlikely that any one tiny volume will be > imaged repeatedly. We've tried several iterations of an aqueous > solution of label, but sealing the prep does not prevent eventual > evaporation. Plus, there seems to be some modification of FITC, for > example, in solution over time (a month) regardless of exposure to > light. I'm not sure I can trust the consistency of colored plastic > slides, since they are not manufactured with this role in mind. One > possibility is uranyl glass slides, but Corning says they don't make > these any more. Does anyone know a source? We're also considering > quantum dots in something like immersion oil , in a sealed chamber > but I don't have any experience with these labels in a hydrophobic > environment. > > Thanks for your insights. > C > > > Carl A. Boswell, Ph.D. > Molecular and Cellular Biology > Univ. of Arizona > 520-954-7053 > FAX 520-621-3709 > > |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I use the Chroma slides and fluorescein solution at known concentrations. Use a water dipping lens with the solution and image away from either the bottom of the dish or the surface of the water. You want to be about middle depth. It won't bleach and as long as you weight it out carefully before putting it into solution you can repeatably create it. Craig On Fri, Jan 14, 2011 at 5:39 PM, Mark Cannell <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > How about Fluorescent plastic slides -I got some free from Chroma. > > Cheers > > > > On 15/01/2011, at 12:34 PM, Boswell, Carl A - (cboswell) wrote: > > Hi all, >> I'm looking for some trustworthy, permanent standard(s) to use in a MP >> system to monitor the condition of detectors and the system in general. >> The goal is a fluorescent source that is immutable. This could done either >> by having unrestricted replacement of bleached label with unbleached label >> (i.e. a solution), or a uniform and very large solid, so that it is unlikely >> that any one tiny volume will be imaged repeatedly. We've tried several >> iterations of an aqueous solution of label, but sealing the prep does not >> prevent eventual evaporation. Plus, there seems to be some modification of >> FITC, for example, in solution over time (a month) regardless of exposure >> to light. I'm not sure I can trust the consistency of colored plastic >> slides, since they are not manufactured with this role in mind. One >> possibility is uranyl glass slides, but Corning says they don't make these >> any more. Does anyone know a source? We're also considering quantum dots >> in something like immersion oil , in a sealed chamber but I don't have any >> experience with these labels in a hydrophobic environment. >> >> Thanks for your insights. >> C >> >> >> Carl A. Boswell, Ph.D. >> Molecular and Cellular Biology >> Univ. of Arizona >> 520-954-7053 >> FAX 520-621-3709 >> >> >> |
|
|
In reply to this post by Mark Cannell
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Fluorescent plastic slides are not stabile. Depending on their exposure to light they bleach or change color. Using a pulsed laser is a good way to make little burn or other chemical reaction spots in them. Here are two old web pages with some of the problems (and benefits!) explained: http://www.einstein.yu.edu/aif/instructions/fluor-ref-slides/01.htm http://www.einstein.yu.edu/aif/instructions/zeiss/liveduo/Zaxis_bleachingtest/index.htm _______________________________________ Michael Cammer, Assistant Research Scientist Skirball Institute of Biomolecular Medicine Lab: (212) 263-3208 Cell: (914) 309-3270 ________________________________________ From: Confocal Microscopy List [[hidden email]] On Behalf Of Mark Cannell [[hidden email]] Sent: Friday, January 14, 2011 7:39 PM To: [hidden email] Subject: Re: 2P standard ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** How about Fluorescent plastic slides -I got some free from Chroma. Cheers On 15/01/2011, at 12:34 PM, Boswell, Carl A - (cboswell) wrote: > Hi all, > I'm looking for some trustworthy, permanent standard(s) to use in a > MP system to monitor the condition of detectors and the system in > general. The goal is a fluorescent source that is immutable. This > could done either by having unrestricted replacement of bleached > label with unbleached label (i.e. a solution), or a uniform and very > large solid, so that it is unlikely that any one tiny volume will be > imaged repeatedly. We've tried several iterations of an aqueous > solution of label, but sealing the prep does not prevent eventual > evaporation. Plus, there seems to be some modification of FITC, for > example, in solution over time (a month) regardless of exposure to > light. I'm not sure I can trust the consistency of colored plastic > slides, since they are not manufactured with this role in mind. One > possibility is uranyl glass slides, but Corning says they don't make > these any more. Does anyone know a source? We're also considering > quantum dots in something like immersion oil , in a sealed chamber > but I don't have any experience with these labels in a hydrophobic > environment. > > Thanks for your insights. > C > > > Carl A. Boswell, Ph.D. > Molecular and Cellular Biology > Univ. of Arizona > 520-954-7053 > FAX 520-621-3709 > > ------------------------------------------------------------ This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email. ================================= |
|
|
In reply to this post by Craig Brideau
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We use fluorescent solutions. We keep a stock in ethanol in the refrigerator, and dilute it to water before use. We measure the absorbance of the final solution in a spectrophotometer evertytime. A droplet on the cover slip is enough to get counts. Sudipta On Fri, 14 Jan 2011 18:05:02 -0700, Craig Brideau wrote > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I use the Chroma slides and fluorescein solution at known concentrations. > Use a water dipping lens with the solution and image away from > either the bottom of the dish or the surface of the water. You want > to be about middle depth. It won't bleach and as long as you weight > it out carefully before putting it into solution you can repeatably > create it. > > Craig > > On Fri, Jan 14, 2011 at 5:39 PM, Mark Cannell > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > How about Fluorescent plastic slides -I got some free from Chroma. > > > > Cheers > > > > > > > > On 15/01/2011, at 12:34 PM, Boswell, Carl A - (cboswell) wrote: > > > > Hi all, > >> I'm looking for some trustworthy, permanent standard(s) to use in a MP > >> system to monitor the condition of detectors and the system in general. > >> The goal is a fluorescent source that is immutable. This could done > >> by having unrestricted replacement of bleached label with unbleached label > >> (i.e. a solution), or a uniform and very large solid, so that it is unlikely > >> that any one tiny volume will be imaged repeatedly. We've tried several > >> iterations of an aqueous solution of label, but sealing the prep does not > >> prevent eventual evaporation. Plus, there seems to be some modification of > >> FITC, for example, in solution over time (a month) regardless of exposure > >> to light. I'm not sure I can trust the consistency of colored plastic > >> slides, since they are not manufactured with this role in mind. One > >> possibility is uranyl glass slides, but Corning says they don't make these > >> any more. Does anyone know a source? We're also considering quantum dots > >> in something like immersion oil , in a sealed chamber but I don't have any > >> experience with these labels in a hydrophobic environment. > >> > >> Thanks for your insights. > >> C > >> > >> > >> Carl A. Boswell, Ph.D. > >> Molecular and Cellular Biology > >> Univ. of Arizona > >> 520-954-7053 > >> FAX 520-621-3709 > >> > >> > >> Dr. Sudipta Maiti Associate Professor Dept. of Chemical Sciences Tata Institute of Fundamental Research Homi Bhabha Raod, Colaba, Mumbai 400005 Ph. 91-22-2278-2716 / 2539 Fax: 91-22-2280-4610 alternate e-mail: [hidden email] url: biophotonics.wetpaint.com |
|
|
In reply to this post by Boswell, Carl A - (cboswell)
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Carl, Instead of using a bleachable fluorescent target, use something that makes great second harmonic generation (SHG). If you MP laser tunes from 690-1040 nm an SHG specimen will be usable from 395-520 nm (maybe you can convince NIH's NCRR to give all of us S10 money to add an OPO or OPA to our MP systems to extend the wavelength range). Besides sugar crystals, starch granules, collagen fibers (see Rich Cole's post), Scott Fraser's lab published a PNAS paper in 2010 using nanoparticles for multicolor SHG-immunofluorescence (antibody conjugation procedure developed by a different Caltech lab and cited by Fraser and colleagues). They used specific types of particles from http://www.nanoamor.com/products Multicolor SHG worked because some types of nanoparticles perferentially underwent SHG in specific excitation wavelength range, whereas others had broad excitation range. Enjoy, George On 1/14/2011 6:34 PM, Boswell, Carl A - (cboswell) wrote: > Hi all, > I'm looking for some trustworthy, permanent standard(s) to use in a MP system to monitor the condition of detectors and the system in general. The goal is a fluorescent source that is immutable. This could done either by having unrestricted replacement of bleached label with unbleached label (i.e. a solution), or a uniform and very large solid, so that it is unlikely that any one tiny volume will be imaged repeatedly. We've tried several iterations of an aqueous solution of label, but sealing the prep does not prevent eventual evaporation. Plus, there seems to be some modification of FITC, for example, in solution over time (a month) regardless of exposure to light. I'm not sure I can trust the consistency of colored plastic slides, since they are not manufactured with this role in mind. One possibility is uranyl glass slides, but Corning says they don't make these any more. Does anyone know a source? We're also considering quantum dots in something like immersion oil , in a sealed chamber but I don't have any experience with these labels in a hydrophobic environment. > > Thanks for your insights. > C > > > Carl A. Boswell, Ph.D. > Molecular and Cellular Biology > Univ. of Arizona > 520-954-7053 > FAX 520-621-3709 > > > |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Since SHG is highly dependent on phase matching how do you ensure that the sample has exactly the same orientation and control the backscatter needed to test the PMTs? Further to an earlier post, my 4 year old plastic plastic slides look as good as new -but then I've looked after them and not exposed them to sunlight. Cheers On 17/01/2011, at 3:38 AM, George McNamara wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi Carl, > > Instead of using a bleachable fluorescent target, use something that > makes great second harmonic generation (SHG). If you MP laser tunes > from 690-1040 nm an SHG specimen will be usable from 395-520 nm > (maybe you can convince NIH's NCRR to give all of us S10 money to > add an OPO or OPA to our MP systems to extend the wavelength range). > > Besides sugar crystals, starch granules, collagen fibers (see Rich > Cole's post), Scott Fraser's lab published a PNAS paper in 2010 > using nanoparticles for multicolor SHG-immunofluorescence (antibody > conjugation procedure developed by a different Caltech lab and cited > by Fraser and colleagues). They used specific types of particles > from http://www.nanoamor.com/products > Multicolor SHG worked because some types of nanoparticles > perferentially underwent SHG in specific excitation wavelength > range, whereas others had broad excitation range. > > Enjoy, > > George > > > > On 1/14/2011 6:34 PM, Boswell, Carl A - (cboswell) wrote: >> Hi all, >> I'm looking for some trustworthy, permanent standard(s) to use in a >> MP system to monitor the condition of detectors and the system in >> general. The goal is a fluorescent source that is immutable. >> This could done either by having unrestricted replacement of >> bleached label with unbleached label (i.e. a solution), or a >> uniform and very large solid, so that it is unlikely that any one >> tiny volume will be imaged repeatedly. We've tried several >> iterations of an aqueous solution of label, but sealing the prep >> does not prevent eventual evaporation. Plus, there seems to be >> some modification of FITC, for example, in solution over time (a >> month) regardless of exposure to light. I'm not sure I can trust >> the consistency of colored plastic slides, since they are not >> manufactured with this role in mind. One possibility is uranyl >> glass slides, but Corning says they don't make these any more. >> Does anyone know a source? We're also considering quantum dots in >> something like immersion oil , in a sealed chamber but I don't have >> any experience with these labels in a hydrophobic environment. >> >> Thanks for your insights. >> C >> >> >> Carl A. Boswell, Ph.D. >> Molecular and Cellular Biology >> Univ. of Arizona >> 520-954-7053 >> FAX 520-621-3709 >> >> >> |
|
|
In reply to this post by George McNamara
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** On 16.01.2011 15:38, George McNamara wrote: > Hi Carl, > > Instead of using a bleachable fluorescent target, use something that > makes great second harmonic generation (SHG). If you MP laser tunes from > 690-1040 nm an SHG specimen will be usable from 395-520 nm (maybe you > can convince NIH's NCRR to give all of us S10 money to add an OPO or OPA > to our MP systems to extend the wavelength range). If you succeed in getting that OPO, you also can use it as a source for THG (Third harmonic generation). A water-glass interface (water on glass slide) will give you a bright and uniform signal if the system is well aligned. It still will work only with your blue detectors, though (e.g. 1270 nm / 3 = 423 nm) Steffen -- ------------------------------------------------------------ Steffen Dietzel, PD Dr. rer. nat Ludwig-Maximilians-Universität München Walter-Brendel-Zentrum für experimentelle Medizin (WBex) Head of light microscopy Mail room: Marchioninistr. 15, D-81377 München Building location: Marchioninistr. 27, München-Großhadern |