40x vs. 60x

classic Classic list List threaded Threaded
♦
♦
4 messages Options
Sarah Kefayati Sarah Kefayati
Reply | Threaded
Open this post in threaded view
♦
♦
|

40x vs. 60x

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello all,
 
I have checked the intensity loss by imaging a fluorescent plastic block with both LUMPlanFI/IR 40x/.8 W and UplanApo/IR 60x/1.2 W through Z axis.The drop of 60x is the way faster that 40x.As I know with the higher NA I should have better light collection.
 
My question is:
 
Could it be because of the fact of higher NA higher sensitivity to the mismatch and higher SA?
or these two lenses are not comparable as 40x is a dipping lens? 
 
Any information is highly appreciated.Thanks
 
Sarah
Ignatius, Mike Ignatius, Mike
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 40x vs. 60x

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Brightness = 10 to the 4th x (Numerical Aperture squared/Magnification squared) 
 
So far three lens:
 
63x/1.4NA = 9.65
100x/1.4NA = 3.8
40x/1.3 NA = 18.9  thus 4x brighter than 100x and twice that of 60x.
 
From Inoue and Spring, Video Microscopy, The Fundamentals, 2nd Edition page 137, 138
 
Mike Ignatius
 
Molecular Probes/Invitrogen
 
 
 


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Sarah Kefayati
Sent: Friday, May 09, 2008 2:55 PM
To: [hidden email]
Subject: 40x vs. 60x

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello all,
 
I have checked the intensity loss by imaging a fluorescent plastic block with both LUMPlanFI/IR 40x/.8 W and UplanApo/IR 60x/1.2 W through Z axis.The drop of 60x is the way faster that 40x.As I know with the higher NA I should have better light collection.
 
My question is:
 
Could it be because of the fact of higher NA higher sensitivity to the mismatch and higher SA?
or these two lenses are not comparable as 40x is a dipping lens? 
 
Any information is highly appreciated.Thanks
 
Sarah
Julio Vazquez Julio Vazquez
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 40x vs. 60x

In reply to this post by Sarah Kefayati
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
=
Sarah, 

your 40x/0.8 and 60x/1.2 are likely to be about the same brightness under ideal conditions (see Mike's post). On the other hand, the 40x/0.8 dipping lens is likely to be more resistant against spherical aberrations as you image deeper into the sample (it is designed to image through a couple of millimeters of water anyway, so a few microns more or less won't make much difference), while the  60x/1.2 will be much more sensitive to spherical aberration (and loss of luminosity), because it is corrected to get the optimal image near the coverslip surface.  If you are imaging deeper than that, you may improve your results by adjusting the correction collar that is probably present on that type of lens, but if you are imaging a thick slice, then you can't avoid SA at all positions, unless perhaps you get a spherical aberration correction device from 3I (http://www.intelligent-imaging.com/spherical/main.php). I have no special interest, and don't know how well it works, but the article referenced (and linked to) in that page may be worth reading.

--
Julio Vazquez
Fred Hutchinson Cancer Research Center
Seattle, WA 98109-1024





On May 9, 2008, at 2:55 PM, Sarah Kefayati wrote:

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello all,
 
I have checked the intensity loss by imaging a fluorescent plastic block with both LUMPlanFI/IR 40x/.8 W and UplanApo/IR 60x/1.2 W through Z axis.The drop of 60x is the way faster that 40x.As I know with the higher NA I should have better light collection.
 
My question is:
 
Could it be because of the fact of higher NA higher sensitivity to the mismatch and higher SA?
or these two lenses are not comparable as 40x is a dipping lens? 
 
Any information is highly appreciated.Thanks
 
Sarah

James Pawley James Pawley
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: 40x vs. 60x

In reply to this post by Sarah Kefayati
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>Brightness = 10 to the 4th x (Numerical
>Aperture squared/Magnification squared)
>
>So far three lens:
>
>63x/1.4NA = 9.65
>100x/1.4NA = 3.8
>40x/1.3 NA = 18.9  thus 4x brighter than 100x and twice that of 60x.
>
>From Inoue and Spring, Video Microscopy, The Fundamentals, 2nd
>Edition page 137, 138
>
>Mike Ignatius
>
>Molecular Probes/Invitrogen


Hi all,

I think that Mike quotes Inoue and Spring correctly. (My copies of
this book are at the UBC Course.)

However, some sources quote a the fourth power NA dependence and this
is really only warranted if the image of the illumination source that
should be present in the BFP of the objective is of sufficient size
to fill the BFP. This is not always true. In addition, it really only
refers to the "apparent brightness to the eye", rather than some
overarching "measure of optical goodness." Whatever you choose, I
sometimes feel that these equations are a bit misleading as they seem
to have led some to conclude that lower mag must be better for
fluorescence. Period..

At a fixed NA, the diameter of the BFP is inversely proportional to
the objective magnification. Once you know the size and shape of the
image of the source formed by the illumination system in the BFP, you
can figure out what fraction of it will pass through the BFP of each
lens. (On an inverted scope with a dry lens, you can see the relation
between the source image and the lens NA  by holding a piece of lens
tissue a few cm above the lens. If the illumination is
Kohler-aligned, you you should see an image of the arc. The dark
circular border represents the rated NA of the objective.).

More than 85% of any light passing the BFP will pass the lens and
illuminate the object. However, in the same way that the area of the
BFP (and hence the amount of light passing through it) is inversely
proportional to the square of the magnification, then the area of the
specimen that is illuminated is also inversely proportional to the
square of the magnification. So assuming that the same image of the
arc is present in the BFP of both 40x and 100x lenses, the photons
striking each square micrometer of the specimen will be the same.

Assuming a uniformly-fluorescent specimen, the amount of signal that
makes it back to a given pixel of the CCD camera, will be inversely
proportional to the objective magnification. So superficially, the
40x will be about 6x brighter than the 100x.

However, when we are recording a digital image, this is not the end
of the story. Nyquist sampling theory says that we want the
dimensions of each pixel of the CCD (referred to the specimen plane)
to be about 2.5x smaller than the resolution limit of the optical
system. This means that, for a given NA (and wavelength),  there
exists a "best" total magnification from the specimen to the plane of
the CCD, and if your 40x lens doesn't provide enough then you should
increase the magnification of the camera-coupling  lens until Nyquist
is happy. Alternatively, if your CCD pixels are two small, you can
"bin" them 2x or 3x to make effective pixels that are 2 or 3 times
larger in linear pixel dimensions.

In any case, as long as the CCD pixels are the same size, when
referred to the specimen, the NA is the same, and the image of the
arc fills the BFP of the lower mag objective fairly uniformly, then
signal recorded in one pixel of the CCD will be the same. (i.e., the
objective mag is irrelevant because the total mag is fixed by
Nyquist.)

So why would you choose one over another? The 4Ox will look brighter
"by eye" although not necessarily 38x brighter or even 6x brighter,
because the image of the arc in the BFP usually not the optimal size.
The 40x will also have a larger field of view making it easier to
find the "good" part of the specimen. On the other hand, the image
will be smaller and harder to "see" and you will bleach a larger area
of the specimen while you are scanning around. Finally, if you fail
to use the field diaphragm to limit the illumination to include only
that part that is imaged by the CCD, you will have a lot more light
reflecting around off optical surfaces,  contributing to background
haze and reducing image contrast.

In addition, as it is much more difficult to correct all of the many
aberrations over the larger field of view of the 40x, it will
probably only have "diffraction-limited" performance over the central
area. By the same token, there may be fewer optical compromises in
the 100x objective, although the 2.5x shorter focal length may impose
limits on the working distance.

Many people will base their choice on whichever lens comes closest to
Nyuquist sampling with their favorite CCD. (Cameras based on the
popular SONY chips have 6.7 micrometer pixels, while those using the
E2V 512x512 EM-CCD chip, preferred by many for disk scanning and
TIRF, has 15 micrometer pixels.)

All of this says nothing about spherical aberration that occurs when
one focuses a dry or an oil lens into fluorescent plastic etc.

Cheers,

Jim Pawley

               **********************************************
Prof. James B. Pawley,                          Ph.  608-263-3147
Room 223, Zoology Research Building,              
FAX  608-265-5315
1117 Johnson Ave., Madison, WI, 53706  
[hidden email]
3D Microscopy of Living Cells Course, June 14-26, 2008, UBC, Vancouver Canada
Info: http://www.3dcourse.ubc.ca/       Applications still being
accepted for Waitlist
               "If it ain't diffraction, it must be statistics." Anon.

>From: Confocal Microscopy List
>[mailto:[hidden email]] On Behalf Of Sarah Kefayati
>Sent: Friday, May 09, 2008 2:55 PM
>To: [hidden email]
>Subject: 40x vs. 60x
>
>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>Hello all,
>
>I have checked the intensity loss by imaging a fluorescent plastic
>block with both LUMPlanFI/IR 40x/.8 W and UplanApo/IR 60x/1.2 W
>through Z axis.The drop of 60x is the way faster that 40x.As I know
>with the higher NA I should have better light collection.
>
>My question is:
>
>Could it be because of the fact of higher NA higher sensitivity to
>the mismatch and higher SA?
>or these two lenses are not comparable as 40x is a dipping lens?
>
>Any information is highly appreciated.Thanks
>
>Sarah


--
--