|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** OK, having slept on it, I now feel that just maybe I can explain what this is all about. If only the list would let us include pictures it would be much easier! Let's assume we have a digital image, from any source, consisting of pixels with a spacing s. The smallest spacing we can resolve in this image is 2s, and this will correspond, in frequency space, with a frequency f. f represents the bandpass limit of this system, no higher frequencies can be passed. Now imagine we have a row of pixels containing the following values: 255 0 255 0 255 0 255 0 255 If we represent these pixels by little squares, we'll have something like a chessboard. Taking a line along this chessboard will give us a square wave. Now this square wave cannot be represented within the bandpass limit of the system, defined by the frequency f. To represent a square wave we need an infinite series of sine waves f + 3f + 5f +7f ..... To get even a crude approximation to a square wave we need f + 3f - that is a frequency three times higher than the image can contain. In other words, we've introduced a whole series of spurious frequencies into our image that not only were not there to start with, they could not possibly have been there. Does this matter? After all, we know they can't be real. It does matter, because we are talking about a visual representation of our data - that's why we drew the little boxes in the first place. Our eyes are very sensitive to edges* and the edges will take over if we let these frequencies come within the bandwidth of our eyes. We will find it very hard to actually see the finest detail in our picture (defined by 2s, remember) because if we enlarge it enough to see this easily we'll also get the edges created by these spurious frequencies. In everyday terms, the pixellation takes over from the picture. Note that in all this discussion I have not mentioned microscopes, cameras or anything - we are just talking about a digital image from any source. It applies to confocal, widefield, and electron microscopes, telescopes, X-ray images and your holiday snaps. Coming back to the microscopic world, if we oversample to the point where r, our minimum resolved distance, is substantially greater than 2s, we may not need to enlarge to the point where we see the spurious frequencies. This is probably why some contributors to this discussion have advocated considerable levels of oversampling (though they probably didn't realise this, they just knew they got good pictures that way). But oversampling in fluorescence can be very hard on our specimens. "But I'm using a CCD detector so my image is made up of little squares". Yes, you can produce a 'coloured in' picture of your detector that way. I'm assuming the image is actually what you want to see, though, not the detector. *Amusingly, the human eye does the same thing to emphasize edges as computer image processing does - it makes the dark side of the edge darker than it is and the light side lighter. Guy PS. This has doubtless confirmed my reputation among some people as an arrogant bastard. They are probably right, but at least I'm an arrogant bastard who tries to help. It's taken me two hours to write this. |
|
Zac Arrac Atelaz |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Thank you so much ABWTTH. It would take me more than those two hours you gave us!!! You DO help Gabriel OH > Date: Mon, 16 Apr 2012 02:31:47 +0000 > From: [hidden email] > Subject: A pixel is not a little square > To: [hidden email] > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > OK, having slept on it, I now feel that just maybe I can explain what this is all about. If only the list would let us include pictures it would be much easier! > > Let's assume we have a digital image, from any source, consisting of pixels with a spacing s. The smallest spacing we can resolve in this image is 2s, and this will correspond, in frequency space, with a frequency f. f represents the bandpass limit of this system, no higher frequencies can be passed. Now imagine we have a row of pixels containing the following values: > > 255 0 255 0 255 0 255 0 255 > > If we represent these pixels by little squares, we'll have something like a chessboard. Taking a line along this chessboard will give us a square wave. Now this square wave cannot be represented within the bandpass limit of the system, defined by the frequency f. To represent a square wave we need an infinite series of sine waves f + 3f + 5f +7f ..... To get even a crude approximation to a square wave we need f + 3f - that is a frequency three times higher than the image can contain. > > In other words, we've introduced a whole series of spurious frequencies into our image that not only were not there to start with, they could not possibly have been there. Does this matter? After all, we know they can't be real. It does matter, because we are talking about a visual representation of our data - that's why we drew the little boxes in the first place. Our eyes are very sensitive to edges* and the edges will take over if we let these frequencies come within the bandwidth of our eyes. We will find it very hard to actually see the finest detail in our picture (defined by 2s, remember) because if we enlarge it enough to see this easily we'll also get the edges created by these spurious frequencies. In everyday terms, the pixellation takes over from the picture. > > Note that in all this discussion I have not mentioned microscopes, cameras or anything - we are just talking about a digital image from any source. It applies to confocal, widefield, and electron microscopes, telescopes, X-ray images and your holiday snaps. Coming back to the microscopic world, if we oversample to the point where r, our minimum resolved distance, is substantially greater than 2s, we may not need to enlarge to the point where we see the spurious frequencies. This is probably why some contributors to this discussion have advocated considerable levels of oversampling (though they probably didn't realise this, they just knew they got good pictures that way). But oversampling in fluorescence can be very hard on our specimens. > > "But I'm using a CCD detector so my image is made up of little squares". Yes, you can produce a 'coloured in' picture of your detector that way. I'm assuming the image is actually what you want to see, though, not the detector. > > *Amusingly, the human eye does the same thing to emphasize edges as computer image processing does - it makes the dark side of the edge darker than it is and the light side lighter. > > Guy > > PS. This has doubtless confirmed my reputation among some people as an arrogant bastard. They are probably right, but at least I'm an arrogant bastard who tries to help. It's taken me two hours to write this. |
|
Mark Cannell |
|
|
In reply to this post by Guy Cox-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I think I see the problem, the spurious frequencies arise from your thinking the _data_ is a continuous function and treating it as such (by "drawing a line ..."), but it is not, it is discrete and can be faithfully represented by a _discrete_ Fourier transform (which folds at Fs/2). The hiighest frequency in the DFT is Fs, but we know we shouldn't look at that right? There are _no_ 'higher harmonics' present in the data, only in ones 'artistic' interpretation for display purposes. If it looks jagged, that is because in reality sampled data really is! The problem really arises because you do not know how to fill in the space between data samples. You can interpolate (or not). If you interpolate you are making a statement about the model underlying the data and have just carried out a fitting exercise. Fitting is NOT raw data presentation. If you just plot data values you make no assumption about what should join the data, no model has been fit to the data. Every scientist should know the difference between a histogram and a continuous distribution and not be fooled by the vertical lines at the histogram boundaries (which is what you show in a pixel image). The choice is yours, in one case you faithfully show unadulterated sampled data (the histogram looks less 'pretty' than a curve) or you fit a model and interpolate. The trouble with the latter is that the model is probably wrong and you hide the defects in the data (e.g. camera pixel size) from the keen eyed reviewer... Of course if the data points are really close together, the myopic reviewer can't see defects in you data :-) ! From Guy's reasoning, it would be impossible to represent any digitally sampled data because you are always pixelating a continuous function (all pictures get mad up of little squares -the printer dumps blobs of ink etc). So, where does the pixelation become acceptable? This is now aesthetic and has nothing to do with science or mathematics (those with perfect vision will always see discretization 'artifacts' more easily) . Cheers Mark On 16/04/2012, at 3:31 AM, Guy Cox wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > OK, having slept on it, I now feel that just maybe I can explain what this is all about. If only the list would let us include pictures it would be much easier! > > Let's assume we have a digital image, from any source, consisting of pixels with a spacing s. The smallest spacing we can resolve in this image is 2s, and this will correspond, in frequency space, with a frequency f. f represents the bandpass limit of this system, no higher frequencies can be passed. Now imagine we have a row of pixels containing the following values: > > 255 0 255 0 255 0 255 0 255 > > If we represent these pixels by little squares, we'll have something like a chessboard. Taking a line along this chessboard will give us a square wave. Now this square wave cannot be represented within the bandpass limit of the system, defined by the frequency f. To represent a square wave we need an infinite series of sine waves f + 3f + 5f +7f ..... To get even a crude approximation to a square wave we need f + 3f - that is a frequency three times higher than the image can contain. > > In other words, we've introduced a whole series of spurious frequencies into our image that not only were not there to start with, they could not possibly have been there. Does this matter? After all, we know they can't be real. It does matter, because we are talking about a visual representation of our data - that's why we drew the little boxes in the first place. Our eyes are very sensitive to edges* and the edges will take over if we let these frequencies come within the bandwidth of our eyes. We will find it very hard to actually see the finest detail in our picture (defined by 2s, remember) because if we enlarge it enough to see this easily we'll also get the edges created by these spurious frequencies. In everyday terms, the pixellation takes over from the picture. > > Note that in all this discussion I have not mentioned microscopes, cameras or anything - we are just talking about a digital image from any source. It applies to confocal, widefield, and electron microscopes, telescopes, X-ray images and your holiday snaps. Coming back to the microscopic world, if we oversample to the point where r, our minimum resolved distance, is substantially greater than 2s, we may not need to enlarge to the point where we see the spurious frequencies. This is probably why some contributors to this discussion have advocated considerable levels of oversampling (though they probably didn't realise this, they just knew they got good pictures that way). But oversampling in fluorescence can be very hard on our specimens. > > "But I'm using a CCD detector so my image is made up of little squares". Yes, you can produce a 'coloured in' picture of your detector that way. I'm assuming the image is actually what you want to see, though, not the detector. > > *Amusingly, the human eye does the same thing to emphasize edges as computer image processing does - it makes the dark side of the edge darker than it is and the light side lighter. > > Guy > > PS. This has doubtless confirmed my reputation among some people as an arrogant bastard. They are probably right, but at least I'm an arrogant bastard who tries to help. It's taken me two hours to write this. |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** " There are _no_ 'higher harmonics' present in the data, only in ones 'artistic' interpretation for display purposes." That is exactly what I said! I also never said that the data is a continuous function, I said it is a series of discrete samples of a continuous function. So when you choose to display it you have to do something. Drawing little boxes is NOT 'doing nothing' and neither is it 'displaying the raw data'. On the contrary - it is corrupting the data with frequencies which shouldn't be there AND confusing the human eye (for which, presumably, we are doing the drawing). The raw numbers are useful - indeed essential - for the computer but fundamentally cannot just 'be displayed' to the human eye as an image. Our sampling rationale is based on sine-wave frequencies and therefore, as Alvy Ray Smith said, sinusoidal mapping is the truest (not the most aesthetic, though this is also true) way of displaying the data. It doesn't add any spurious higher harmonics, it presents the data as accurately as our sampling permits. Drawing little boxes may be easier, but it is just as much mapping the measured samples to a displayed image - the difference is that this method is both inaccurate and un-aesthetic. If your CD player spat out square waves to the speakers, you'd take it back to the shop pretty promptly! Guy -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mark Cannell Sent: Monday, 16 April 2012 5:27 PM To: [hidden email] Subject: Re: A pixel is not a little square ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I think I see the problem, the spurious frequencies arise from your thinking the _data_ is a continuous function and treating it as such (by "drawing a line ..."), but it is not, it is discrete and can be faithfully represented by a _discrete_ Fourier transform (which folds at Fs/2). The hiighest frequency in the DFT is Fs, but we know we shouldn't look at that right? There are _no_ 'higher harmonics' present in the data, only in ones 'artistic' interpretation for display purposes. If it looks jagged, that is because in reality sampled data really is! The problem really arises because you do not know how to fill in the space between data samples. You can interpolate (or not). If you interpolate you are making a statement about the model underlying the data and have just carried out a fitting exercise. Fitting is NOT raw data presentation. If you just plot data values you make no assumption about what should join the data, no model has been fit to the data. Every scientist should know the difference between a histogram and a continuous distribution and not be fooled by the vertical lines at the histogram boundaries (which is what you show in a pixel image). The choice is yours, in one case you faithfully show unadulterated sampled data (the histogram looks less 'pretty' than a curve) or you fit a model and interpolate. The trouble with the latter is that the model is probably wrong and you hide the defects in the data (e.g. camera pixel size) from the keen eyed reviewer... Of course if the data points are really close together, the myopic reviewer can't see defects in you data :-) ! From Guy's reasoning, it would be impossible to represent any digitally sampled data because you are always pixelating a continuous function (all pictures get mad up of little squares -the printer dumps blobs of ink etc). So, where does the pixelation become acceptable? This is now aesthetic and has nothing to do with science or mathematics (those with perfect vision will always see discretization 'artifacts' more easily) . Cheers Mark On 16/04/2012, at 3:31 AM, Guy Cox wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > OK, having slept on it, I now feel that just maybe I can explain what this is all about. If only the list would let us include pictures it would be much easier! > > Let's assume we have a digital image, from any source, consisting of pixels with a spacing s. The smallest spacing we can resolve in this image is 2s, and this will correspond, in frequency space, with a frequency f. f represents the bandpass limit of this system, no higher frequencies can be passed. Now imagine we have a row of pixels containing the following values: > > 255 0 255 0 255 0 255 0 255 > > If we represent these pixels by little squares, we'll have something like a chessboard. Taking a line along this chessboard will give us a square wave. Now this square wave cannot be represented within the bandpass limit of the system, defined by the frequency f. To represent a square wave we need an infinite series of sine waves f + 3f + 5f +7f ..... To get even a crude approximation to a square wave we need f + 3f - that is a frequency three times higher than the image can contain. > > In other words, we've introduced a whole series of spurious frequencies into our image that not only were not there to start with, they could not possibly have been there. Does this matter? After all, we know they can't be real. It does matter, because we are talking about a visual representation of our data - that's why we drew the little boxes in the first place. Our eyes are very sensitive to edges* and the edges will take over if we let these frequencies come within the bandwidth of our eyes. We will find it very hard to actually see the finest detail in our picture (defined by 2s, remember) because if we enlarge it enough to see this easily we'll also get the edges created by these spurious frequencies. In everyday terms, the pixellation takes over from the picture. > > Note that in all this discussion I have not mentioned microscopes, cameras or anything - we are just talking about a digital image from any source. It applies to confocal, widefield, and electron microscopes, telescopes, X-ray images and your holiday snaps. Coming back to the microscopic world, if we oversample to the point where r, our minimum resolved distance, is substantially greater than 2s, we may not need to enlarge to the point where we see the spurious frequencies. This is probably why some contributors to this discussion have advocated considerable levels of oversampling (though they probably didn't realise this, they just knew they got good pictures that way). But oversampling in fluorescence can be very hard on our specimens. > > "But I'm using a CCD detector so my image is made up of little squares". Yes, you can produce a 'coloured in' picture of your detector that way. I'm assuming the image is actually what you want to see, though, not the detector. > > *Amusingly, the human eye does the same thing to emphasize edges as computer image processing does - it makes the dark side of the edge darker than it is and the light side lighter. > > Guy > > PS. This has doubtless confirmed my reputation among some people as an arrogant bastard. They are probably right, but at least I'm an arrogant bastard who tries to help. It's taken me two hours to write this. |
|
Mark Cannell |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Sorry Guy, I still think you don't see the point I'm trying to make. The camera actually says "The mean signal from x to x+dx is ..." (where dx is the sensor pixel size). It does NOT say the signal at x is 'K' and that is where I think the confusion lies. The camera output is a 2D 'histogram' and showing little boxes with the same intensity is (I say again) a perfectly accurate representation of the data ( i.e. F(x) for x -> x + dx = K). With respect, it is not, as you say inaccurate -even if it is unaesthetic. If you fit a sinusoid you have just carried out a fitting exercise... That is not a "more accurate" presentation of the data despite what your Smith says (even if it may be a more accurate representation of the object which has been discretized). One should not loose sight of the fact that you have made some (possibly large) assumptions in the fitting process. Put mathematically, if you smooth out the displayed pixel edges you extend the actual sampling frequency (note how you are putting new unrecorded samples between recorded data values -which is what drawing a line between points actually does) -you are adding information to the data that was NOT present in the RAW data. It may be that your additional information is correct and adds value (e.g. the band limit of the microscope is...) but one should not loose sight of distinction between the addition of data/information by the experimenter (which may or may not be wrong) and that reported by the instrument (the closest to truth the experimenter can get). At the risk of boring some readers on this list, let me emphasize my point : The camera actually says "The mean signal from x to x+dx is ..." (where dx is the sensor pixel size). It does NOT say the signal at x is 'K' . This can be portrayed as a square with constant color and I can think of no other truer portrayal of the measured data. Hopefully dx is less than the resolution of the viewer at final display resolution but if it is not, then the only choice (IMHO) is between aesthetics (or some other goal) and truthfully displaying the recorded data -there is no middle ground. Cheers Mark PS My CD player can't output square waves because the detector etc. has a rather finite bandwidth... Even if it could, my ears are too many dB down at 44 kHz to sample it correctly and hear the artifacts introduced by digital sampling ... :-) On 16/04/2012, at 9:18 AM, Guy Cox wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > " There are _no_ 'higher harmonics' present in the data, only in ones 'artistic' interpretation for display purposes." That is exactly what I said! > > I also never said that the data is a continuous function, I said it is a series of discrete samples of a continuous function. So when you choose to display it you have to do something. Drawing little boxes is NOT 'doing nothing' and neither is it 'displaying the raw data'. On the contrary - it is corrupting the data with frequencies which shouldn't be there AND confusing the human eye (for which, presumably, we are doing the drawing). The raw numbers are useful - indeed essential - for the computer but fundamentally cannot just 'be displayed' to the human eye as an image. Our sampling rationale is based on sine-wave frequencies and therefore, as Alvy Ray Smith said, sinusoidal mapping is the truest (not the most aesthetic, though this is also true) way of displaying the data. It doesn't add any spurious higher harmonics, it presents the data as accurately as our sampling permits. Drawing little boxes may be easier, but it is just as much mapping the measured samples to a displayed image - the difference is that this method is both inaccurate and un-aesthetic. > > If your CD player spat out square waves to the speakers, you'd take it back to the shop pretty promptly! > > Guy > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mark Cannell > Sent: Monday, 16 April 2012 5:27 PM > To: [hidden email] > Subject: Re: A pixel is not a little square > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I think I see the problem, the spurious frequencies arise from your thinking the _data_ is a continuous function and treating it as such (by "drawing a line ..."), but it is not, it is discrete and can be faithfully represented by a _discrete_ Fourier transform (which folds at Fs/2). The hiighest frequency in the DFT is Fs, but we know we shouldn't look at that right? There are _no_ 'higher harmonics' present in the data, only in ones 'artistic' interpretation for display purposes. > > If it looks jagged, that is because in reality sampled data really is! The problem really arises because you do not know how to fill in the space between data samples. You can interpolate (or not). If you interpolate you are making a statement about the model underlying the data and have just carried out a fitting exercise. Fitting is NOT raw data presentation. If you just plot data values you make no assumption about what should join the data, no model has been fit to the data. Every scientist should know the difference between a histogram and a continuous distribution and not be fooled by the vertical lines at the histogram boundaries (which is what you show in a pixel image). > > The choice is yours, in one case you faithfully show unadulterated sampled data (the histogram looks less 'pretty' than a curve) or you fit a model and interpolate. The trouble with the latter is that the model is probably wrong and you hide the defects in the data (e.g. camera pixel size) from the keen eyed reviewer... Of course if the data points are really close together, the myopic reviewer can't see defects in you data :-) ! From Guy's reasoning, it would be impossible to represent any digitally sampled data because you are always pixelating a continuous function (all pictures get mad up of little squares -the printer dumps blobs of ink etc). So, where does the pixelation become acceptable? This is now aesthetic and has nothing to do with science or mathematics (those with perfect vision will always see discretization 'artifacts' more easily) . > > Cheers Mark > > On 16/04/2012, at 3:31 AM, Guy Cox wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> OK, having slept on it, I now feel that just maybe I can explain what this is all about. If only the list would let us include pictures it would be much easier! >> >> Let's assume we have a digital image, from any source, consisting of pixels with a spacing s. The smallest spacing we can resolve in this image is 2s, and this will correspond, in frequency space, with a frequency f. f represents the bandpass limit of this system, no higher frequencies can be passed. Now imagine we have a row of pixels containing the following values: >> >> 255 0 255 0 255 0 255 0 255 >> >> If we represent these pixels by little squares, we'll have something like a chessboard. Taking a line along this chessboard will give us a square wave. Now this square wave cannot be represented within the bandpass limit of the system, defined by the frequency f. To represent a square wave we need an infinite series of sine waves f + 3f + 5f +7f ..... To get even a crude approximation to a square wave we need f + 3f - that is a frequency three times higher than the image can contain. >> >> In other words, we've introduced a whole series of spurious frequencies into our image that not only were not there to start with, they could not possibly have been there. Does this matter? After all, we know they can't be real. It does matter, because we are talking about a visual representation of our data - that's why we drew the little boxes in the first place. Our eyes are very sensitive to edges* and the edges will take over if we let these frequencies come within the bandwidth of our eyes. We will find it very hard to actually see the finest detail in our picture (defined by 2s, remember) because if we enlarge it enough to see this easily we'll also get the edges created by these spurious frequencies. In everyday terms, the pixellation takes over from the picture. >> >> Note that in all this discussion I have not mentioned microscopes, cameras or anything - we are just talking about a digital image from any source. It applies to confocal, widefield, and electron microscopes, telescopes, X-ray images and your holiday snaps. Coming back to the microscopic world, if we oversample to the point where r, our minimum resolved distance, is substantially greater than 2s, we may not need to enlarge to the point where we see the spurious frequencies. This is probably why some contributors to this discussion have advocated considerable levels of oversampling (though they probably didn't realise this, they just knew they got good pictures that way). But oversampling in fluorescence can be very hard on our specimens. >> >> "But I'm using a CCD detector so my image is made up of little squares". Yes, you can produce a 'coloured in' picture of your detector that way. I'm assuming the image is actually what you want to see, though, not the detector. >> >> *Amusingly, the human eye does the same thing to emphasize edges as computer image processing does - it makes the dark side of the edge darker than it is and the light side lighter. >> >> Guy >> >> PS. This has doubtless confirmed my reputation among some people as an arrogant bastard. They are probably right, but at least I'm an arrogant bastard who tries to help. It's taken me two hours to write this. |
|
Jeremy Adler-4 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I am with Guy on pixels. Most comments are spot on about the size of pixels needed for Nyquist sampling. But while sampling provides the data needed to reconstruct the original signal, it (the pixellated image) is not the best reconstruction envisaged by sampling theory. This requires a filtering of the raw data. Look at a sine wave sampled at 2.3 times its frequency - its amplitude is greattly reduced and it requires some imagination to spot the underlying frequency, the sampling is OK but the final step is missing. An additional problem with fluorophores is Poisson noise - take an image of a static object and take a second image - they are not the same because the number of emitted photons varies, even when light source is stable. Poisson noise is unavoidable and can only be minized by increasing the the number of photons detected. Poisson noise accounts for most of the appearance of pixellation seen in raw confocal images - sampling at the Nyquist spacing means that the smallest resolvable object will spread out over several pixels, it is Poisson noise that creates the sudden changes in intensity between adjacent pixels. The solution is image deconvolution, which attempts to deal with both Poisson noise and reconstruct the most representative image from the raw data that has ideally been correctly sampled. Quoting Mark Cannell <[hidden email]>: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Sorry Guy, I still think you don't see the point I'm trying to make. > The camera actually says "The mean signal from x to x+dx is ..." > (where dx is the sensor pixel size). It does NOT say the signal at x > is 'K' and that is where I think the confusion lies. The camera > output is a 2D 'histogram' and showing little boxes with the same > intensity is (I say again) a perfectly accurate representation of > the data ( i.e. F(x) for x -> x + dx = K). With respect, it is not, > as you say inaccurate -even if it is unaesthetic. If you fit a > sinusoid you have just carried out a fitting exercise... That is not > a "more accurate" presentation of the data despite what your Smith > says (even if it may be a more accurate representation of the object > which has been discretized). One should not loose sight of the fact > that you have made some (possibly large) assumptions in the fitting > process. > > Put mathematically, if you smooth out the displayed pixel edges you > extend the actual sampling frequency (note how you are putting new > unrecorded samples between recorded data values -which is what > drawing a line between points actually does) -you are adding > information to the data that was NOT present in the RAW data. It may > be that your additional information is correct and adds value (e.g. > the band limit of the microscope is...) but one should not loose > sight of distinction between the addition of data/information by the > experimenter (which may or may not be wrong) and that reported by > the instrument (the closest to truth the experimenter can get). > > At the risk of boring some readers on this list, let me emphasize my > point : The camera actually says "The mean signal from x to x+dx is > ..." (where dx is the sensor pixel size). It does NOT say the signal > at x is 'K' . This can be portrayed as a square with constant color > and I can think of no other truer portrayal of the measured data. > Hopefully dx is less than the resolution of the viewer at final > display resolution but if it is not, then the only choice (IMHO) is > between aesthetics (or some other goal) and truthfully displaying > the recorded data -there is no middle ground. > > Cheers Mark > > PS My CD player can't output square waves because the detector etc. > has a rather finite bandwidth... Even if it could, my ears are too > many dB down at 44 kHz to sample it correctly and hear the artifacts > introduced by digital sampling ... :-) > > On 16/04/2012, at 9:18 AM, Guy Cox wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> " There are _no_ 'higher harmonics' present in the data, only in >> ones 'artistic' interpretation for display purposes." That is >> exactly what I said! >> >> I also never said that the data is a continuous function, I said it >> is a series of discrete samples of a continuous function. So when >> you choose to display it you have to do something. Drawing little >> boxes is NOT 'doing nothing' and neither is it 'displaying the raw >> data'. On the contrary - it is corrupting the data with >> frequencies which shouldn't be there AND confusing the human eye >> (for which, presumably, we are doing the drawing). The raw numbers >> are useful - indeed essential - for the computer but fundamentally >> cannot just 'be displayed' to the human eye as an image. Our >> sampling rationale is based on sine-wave frequencies and therefore, >> as Alvy Ray Smith said, sinusoidal mapping is the truest (not the >> most aesthetic, though this is also true) way of displaying the >> data. It doesn't add any spurious higher harmonics, it presents >> the data as accurately as our sampling permits. Drawing little >> boxes may be easier, but it is just as much mapping the measured >> samples to a displayed image - the difference is that this method >> is both inaccurate and un-aesthetic. >> >> If your CD player spat out square waves to the speakers, you'd take >> it back to the shop pretty promptly! >> >> >> >> Guy >> >> -----Original Message----- >> From: Confocal Microscopy List >> [mailto:[hidden email]] On Behalf Of Mark Cannell >> Sent: Monday, 16 April 2012 5:27 PM >> To: [hidden email] >> Subject: Re: A pixel is not a little square >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> I think I see the problem, the spurious frequencies arise from your >> thinking the _data_ is a continuous function and treating it as >> such (by "drawing a line ..."), but it is not, it is discrete and >> can be faithfully represented by a _discrete_ Fourier transform >> (which folds at Fs/2). The hiighest frequency in the DFT is Fs, but >> we know we shouldn't look at that right? There are _no_ 'higher >> harmonics' present in the data, only in ones 'artistic' >> interpretation for display purposes. >> >> If it looks jagged, that is because in reality sampled data really >> is! The problem really arises because you do not know how to fill >> in the space between data samples. You can interpolate (or not). >> If you interpolate you are making a statement about the model >> underlying the data and have just carried out a fitting exercise. >> Fitting is NOT raw data presentation. If you just plot data values >> you make no assumption about what should join the data, no model >> has been fit to the data. Every scientist should know the >> difference between a histogram and a continuous distribution and >> not be fooled by the vertical lines at the histogram boundaries >> (which is what you show in a pixel image). >> >> The choice is yours, in one case you faithfully show unadulterated >> sampled data (the histogram looks less 'pretty' than a curve) or >> you fit a model and interpolate. The trouble with the latter is >> that the model is probably wrong and you hide the defects in the >> data (e.g. camera pixel size) from the keen eyed reviewer... Of >> course if the data points are really close together, the myopic >> reviewer can't see defects in you data :-) ! From Guy's >> reasoning, it would be impossible to represent any digitally >> sampled data because you are always pixelating a continuous >> function (all pictures get mad up of little squares -the printer >> dumps blobs of ink etc). So, where does the pixelation become >> acceptable? This is now aesthetic and has nothing to do with >> science or mathematics (those with perfect vision will always see >> discretization 'artifacts' more easily) . >> >> Cheers Mark >> >> On 16/04/2012, at 3:31 AM, Guy Cox wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> OK, having slept on it, I now feel that just maybe I can explain >>> what this is all about. If only the list would let us include >>> pictures it would be much easier! >>> >>> Let's assume we have a digital image, from any source, consisting >>> of pixels with a spacing s. The smallest spacing we can resolve >>> in this image is 2s, and this will correspond, in frequency space, >>> with a frequency f. f represents the bandpass limit of this >>> system, no higher frequencies can be passed. Now imagine we have >>> a row of pixels containing the following values: >>> >>> 255 0 255 0 255 0 255 0 255 >>> >>> If we represent these pixels by little squares, we'll have >>> something like a chessboard. Taking a line along this chessboard >>> will give us a square wave. Now this square wave cannot be >>> represented within the bandpass limit of the system, defined by >>> the frequency f. To represent a square wave we need an infinite >>> series of sine waves f + 3f + 5f +7f ..... To get even a crude >>> approximation to a square wave we need f + 3f - that is a >>> frequency three times higher than the image can contain. >>> >>> In other words, we've introduced a whole series of spurious >>> frequencies into our image that not only were not there to start >>> with, they could not possibly have been there. Does this matter? >>> After all, we know they can't be real. It does matter, because >>> we are talking about a visual representation of our data - that's >>> why we drew the little boxes in the first place. Our eyes are >>> very sensitive to edges* and the edges will take over if we let >>> these frequencies come within the bandwidth of our eyes. We will >>> find it very hard to actually see the finest detail in our picture >>> (defined by 2s, remember) because if we enlarge it enough to see >>> this easily we'll also get the edges created by these spurious >>> frequencies. In everyday terms, the pixellation takes over from >>> the picture. >>> >>> Note that in all this discussion I have not mentioned >>> microscopes, cameras or anything - we are just talking about a >>> digital image from any source. It applies to confocal, widefield, >>> and electron microscopes, telescopes, X-ray images and your >>> holiday snaps. Coming back to the microscopic world, if we >>> oversample to the point where r, our minimum resolved distance, is >>> substantially greater than 2s, we may not need to enlarge to the >>> point where we see the spurious frequencies. This is probably why >>> some contributors to this discussion have advocated considerable >>> levels of oversampling (though they probably didn't realise this, >>> they just knew they got good pictures that way). But oversampling >>> in fluorescence can be very hard on our specimens. >>> >>> "But I'm using a CCD detector so my image is made up of little >>> squares". Yes, you can produce a 'coloured in' picture of your >>> detector that way. I'm assuming the image is actually what you >>> want to see, though, not the detector. >>> >>> *Amusingly, the human eye does the same thing to emphasize edges >>> as computer image processing does - it makes the dark side of the >>> edge darker than it is and the light side lighter. >>> >>> >>> Guy >>> >>> PS. This has doubtless confirmed my reputation among some people >>> as an arrogant bastard. They are probably right, but at least I'm >>> an arrogant bastard who tries to help. It's taken me two hours to >>> write this. > Jeremy Adler IGP Rudbeckslaboratoriet Daghammersköljdsväg 20 751 85 Uppsala Sweden 0046 (0)18 471 4607 |
|
|
In reply to this post by Mark Cannell
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Mark, You are continuing to confuse the samples with the representation of those samples. Let's imagine we have a series of data points: 100 90 80 70 60 50 40 We are mapping these on an image where each is separated by a defined distance. So we need to fill in this distance. You are saying that the 'correct' representation is: 100 100 100 100 100 90 90 90 90 90 80 80 80 80 80 70 70 70 70 60 60 60 60 60 50 50 50 50 50 40 I am saying that this is a wildly implausible and totally unjustified interpretation, and the best representation we can derive from the data is: 100 98 96 94 92 90 88 86 84 82 80 78 76 74 72 70 68 66 64 62 60 58 56 54 52 50 48 46 44 42 40 EITHER way we are interpolating the sampled data - we have no option - so let's just get over this. Your proposed representation includes detail that we could not possibly detect, mine does not. Remember, these are SAMPLES. Neither representation changes our recorded data. End of story, IMHO. How did we get into this mess? Why does everyone then 'do it wrong'? Well, actually, everyone doesn't. Scanning probe microscopes always remap - because by the time they appeared the computing power to do it was available. When confocal microscopes first became widely available, in 1987, the data they produced completely overwhelmed available computing power (believe me, I was there and writing software). So we got used to the 'quick and dirty' approach. Consumer digital cameras do a sort of remap because the Bayer mosaic requires it, but modern sensors so far exceed the resolution of the camera optics that we never get to see any spurious frequencies anyway. Computer games consoles always remap. So do X-ray and EM tomography systems. Guy (arrogant bastard) ----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mark Cannell Sent: Monday, 16 April 2012 7:37 PM To: [hidden email] Subject: Re: A pixel is not a little square ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Sorry Guy, I still think you don't see the point I'm trying to make. The camera actually says "The mean signal from x to x+dx is ..." (where dx is the sensor pixel size). It does NOT say the signal at x is 'K' and that is where I think the confusion lies. The camera output is a 2D 'histogram' and showing little boxes with the same intensity is (I say again) a perfectly accurate representation of the data ( i.e. F(x) for x -> x + dx = K). With respect, it is not, as you say inaccurate -even if it is unaesthetic. If you fit a sinusoid you have just carried out a fitting exercise... That is not a "more accurate" presentation of the data despite what your Smith says (even if it may be a more accurate representation of the object which has been discretized). One should not loose sight of the fact that you have made some (possibly large) assumptions in the fitting process. Put mathematically, if you smooth out the displayed pixel edges you extend the actual sampling frequency (note how you are putting new unrecorded samples between recorded data values -which is what drawing a line between points actually does) -you are adding information to the data that was NOT present in the RAW data. It may be that your additional information is correct and adds value (e.g. the band limit of the microscope is...) but one should not loose sight of distinction between the addition of data/information by the experimenter (which may or may not be wrong) and that reported by the instrument (the closest to truth the experimenter can get). At the risk of boring some readers on this list, let me emphasize my point : The camera actually says "The mean signal from x to x+dx is ..." (where dx is the sensor pixel size). It does NOT say the signal at x is 'K' . This can be portrayed as a square with constant color and I can think of no other truer portrayal of the measured data. Hopefully dx is less than the resolution of the viewer at final display resolution but if it is not, then the only choice (IMHO) is between aesthetics (or some other goal) and truthfully displaying the recorded data -there is no middle ground. Cheers Mark PS My CD player can't output square waves because the detector etc. has a rather finite bandwidth... Even if it could, my ears are too many dB down at 44 kHz to sample it correctly and hear the artifacts introduced by digital sampling ... :-) On 16/04/2012, at 9:18 AM, Guy Cox wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > " There are _no_ 'higher harmonics' present in the data, only in ones 'artistic' interpretation for display purposes." That is exactly what I said! > > I also never said that the data is a continuous function, I said it is a series of discrete samples of a continuous function. So when you choose to display it you have to do something. Drawing little boxes is NOT 'doing nothing' and neither is it 'displaying the raw data'. On the contrary - it is corrupting the data with frequencies which shouldn't be there AND confusing the human eye (for which, presumably, we are doing the drawing). The raw numbers are useful - indeed essential - for the computer but fundamentally cannot just 'be displayed' to the human eye as an image. Our sampling rationale is based on sine-wave frequencies and therefore, as Alvy Ray Smith said, sinusoidal mapping is the truest (not the most aesthetic, though this is also true) way of displaying the data. It doesn't add any spurious higher harmonics, it presents the data as accurately as our sampling permits. Drawing little boxes may be easier, but it is just as much mapping the measured samples to a displayed image - the difference is that this method is both inaccurate and un-aesthetic. > > If your CD player spat out square waves to the speakers, you'd take it back to the shop pretty promptly! > > Guy > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mark Cannell > Sent: Monday, 16 April 2012 5:27 PM > To: [hidden email] > Subject: Re: A pixel is not a little square > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I think I see the problem, the spurious frequencies arise from your thinking the _data_ is a continuous function and treating it as such (by "drawing a line ..."), but it is not, it is discrete and can be faithfully represented by a _discrete_ Fourier transform (which folds at Fs/2). The hiighest frequency in the DFT is Fs, but we know we shouldn't look at that right? There are _no_ 'higher harmonics' present in the data, only in ones 'artistic' interpretation for display purposes. > > If it looks jagged, that is because in reality sampled data really is! The problem really arises because you do not know how to fill in the space between data samples. You can interpolate (or not). If you interpolate you are making a statement about the model underlying the data and have just carried out a fitting exercise. Fitting is NOT raw data presentation. If you just plot data values you make no assumption about what should join the data, no model has been fit to the data. Every scientist should know the difference between a histogram and a continuous distribution and not be fooled by the vertical lines at the histogram boundaries (which is what you show in a pixel image). > > The choice is yours, in one case you faithfully show unadulterated sampled data (the histogram looks less 'pretty' than a curve) or you fit a model and interpolate. The trouble with the latter is that the model is probably wrong and you hide the defects in the data (e.g. camera pixel size) from the keen eyed reviewer... Of course if the data points are really close together, the myopic reviewer can't see defects in you data :-) ! From Guy's reasoning, it would be impossible to represent any digitally sampled data because you are always pixelating a continuous function (all pictures get mad up of little squares -the printer dumps blobs of ink etc). So, where does the pixelation become acceptable? This is now aesthetic and has nothing to do with science or mathematics (those with perfect vision will always see discretization 'artifacts' more easily) . > > Cheers Mark > > On 16/04/2012, at 3:31 AM, Guy Cox wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> OK, having slept on it, I now feel that just maybe I can explain what this is all about. If only the list would let us include pictures it would be much easier! >> >> Let's assume we have a digital image, from any source, consisting of pixels with a spacing s. The smallest spacing we can resolve in this image is 2s, and this will correspond, in frequency space, with a frequency f. f represents the bandpass limit of this system, no higher frequencies can be passed. Now imagine we have a row of pixels containing the following values: >> >> 255 0 255 0 255 0 255 0 255 >> >> If we represent these pixels by little squares, we'll have something like a chessboard. Taking a line along this chessboard will give us a square wave. Now this square wave cannot be represented within the bandpass limit of the system, defined by the frequency f. To represent a square wave we need an infinite series of sine waves f + 3f + 5f +7f ..... To get even a crude approximation to a square wave we need f + 3f - that is a frequency three times higher than the image can contain. >> >> In other words, we've introduced a whole series of spurious frequencies into our image that not only were not there to start with, they could not possibly have been there. Does this matter? After all, we know they can't be real. It does matter, because we are talking about a visual representation of our data - that's why we drew the little boxes in the first place. Our eyes are very sensitive to edges* and the edges will take over if we let these frequencies come within the bandwidth of our eyes. We will find it very hard to actually see the finest detail in our picture (defined by 2s, remember) because if we enlarge it enough to see this easily we'll also get the edges created by these spurious frequencies. In everyday terms, the pixellation takes over from the picture. >> >> Note that in all this discussion I have not mentioned microscopes, cameras or anything - we are just talking about a digital image from any source. It applies to confocal, widefield, and electron microscopes, telescopes, X-ray images and your holiday snaps. Coming back to the microscopic world, if we oversample to the point where r, our minimum resolved distance, is substantially greater than 2s, we may not need to enlarge to the point where we see the spurious frequencies. This is probably why some contributors to this discussion have advocated considerable levels of oversampling (though they probably didn't realise this, they just knew they got good pictures that way). But oversampling in fluorescence can be very hard on our specimens. >> >> "But I'm using a CCD detector so my image is made up of little squares". Yes, you can produce a 'coloured in' picture of your detector that way. I'm assuming the image is actually what you want to see, though, not the detector. >> >> *Amusingly, the human eye does the same thing to emphasize edges as computer image processing does - it makes the dark side of the edge darker than it is and the light side lighter. >> >> Guy >> >> PS. This has doubtless confirmed my reputation among some people as an arrogant bastard. They are probably right, but at least I'm an arrogant bastard who tries to help. It's taken me two hours to write this. |
|
lechristophe |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Guy, But in the CCD case the sampling is integrated over the whole pixel area of the sensor (assuming pixel coverage of 100% which is the case for EMCCDs for exemple). There is no information about where did the photons arrive on the square detector portion corresponding to a pixel. The value X you get as intensity for this pixel is the integrated density over the whole detector portion, so why would you take that value as a point-measurement corresponding to the center of this area ? In that CCD case, isn't the mage indeed an histogram with a bin that has (quite literally) a width equal to the pixel size, not a sample at mathematical points corresponding to the centers of the pixels ? Christophe On Mon, Apr 16, 2012 at 14:36, Guy Cox <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Mark, > > You are continuing to confuse the samples with the > representation of those samples. > > Let's imagine we have a series of data points: 100 90 80 70 60 50 40 > > We are mapping these on an image where each is separated by a defined > distance. So we need to fill in this distance. > > You are saying that the 'correct' representation is: > 100 100 100 100 100 90 90 90 90 90 80 80 80 80 80 > 70 70 70 70 60 60 60 60 60 50 50 50 50 50 40 > > I am saying that this is a wildly implausible and totally unjustified > interpretation, and the best representation we can derive from the data is: > 100 98 96 94 92 90 88 86 84 82 80 78 76 74 72 > 70 68 66 64 62 60 58 56 54 52 50 48 46 44 42 > 40 > > EITHER way we are interpolating the sampled data - we have no option - so > let's just get over this. Your proposed representation includes detail > that we could not possibly detect, mine does not. Remember, these are > SAMPLES. Neither representation changes our recorded data. End of story, > IMHO. > > How did we get into this mess? Why does everyone then 'do it wrong'? > Well, actually, everyone doesn't. Scanning probe microscopes always remap > - because by the time they appeared the computing power to do it was > available. When confocal microscopes first became widely available, in > 1987, the data they produced completely overwhelmed available computing > power (believe me, I was there and writing software). So we got used to > the 'quick and dirty' approach. Consumer digital cameras do a sort of > remap because the Bayer mosaic requires it, but modern sensors so far > exceed the resolution of the camera optics that we never get to see any > spurious frequencies anyway. Computer games consoles always remap. So do > X-ray and EM tomography systems. > > Guy (arrogant bastard) > > > > ----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Mark Cannell > Sent: Monday, 16 April 2012 7:37 PM > To: [hidden email] > Subject: Re: A pixel is not a little square > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Sorry Guy, I still think you don't see the point I'm trying to make. The > camera actually says "The mean signal from x to x+dx is ..." (where dx is > the sensor pixel size). It does NOT say the signal at x is 'K' and that is > where I think the confusion lies. The camera output is a 2D 'histogram' and > showing little boxes with the same intensity is (I say again) a perfectly > accurate representation of the data ( i.e. F(x) for x -> x + dx = K). With > respect, it is not, as you say inaccurate -even if it is unaesthetic. If > you fit a sinusoid you have just carried out a fitting exercise... That is > not a "more accurate" presentation of the data despite what your Smith says > (even if it may be a more accurate representation of the object which has > been discretized). One should not loose sight of the fact that you have > made some (possibly large) assumptions in the fitting process. > > Put mathematically, if you smooth out the displayed pixel edges you extend > the actual sampling frequency (note how you are putting new unrecorded > samples between recorded data values -which is what drawing a line between > points actually does) -you are adding information to the data that was NOT > present in the RAW data. It may be that your additional information is > correct and adds value (e.g. the band limit of the microscope is...) but > one should not loose sight of distinction between the addition of > data/information by the experimenter (which may or may not be wrong) and > that reported by the instrument (the closest to truth the experimenter can > get). > > At the risk of boring some readers on this list, let me emphasize my point > : The camera actually says "The mean signal from x to x+dx is ..." (where > dx is the sensor pixel size). It does NOT say the signal at x is 'K' . This > can be portrayed as a square with constant color and I can think of no > other truer portrayal of the measured data. Hopefully dx is less than the > resolution of the viewer at final display resolution but if it is not, then > the only choice (IMHO) is between aesthetics (or some other goal) and > truthfully displaying the recorded data -there is no middle ground. > > Cheers Mark > > PS My CD player can't output square waves because the detector etc. has a > rather finite bandwidth... Even if it could, my ears are too many dB down > at 44 kHz to sample it correctly and hear the artifacts introduced by > digital sampling ... :-) > > On 16/04/2012, at 9:18 AM, Guy Cox wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > " There are _no_ 'higher harmonics' present in the data, only in ones > 'artistic' interpretation for display purposes." That is exactly what I > said! > > > > I also never said that the data is a continuous function, I said it is a > series of discrete samples of a continuous function. So when you choose to > display it you have to do something. Drawing little boxes is NOT 'doing > nothing' and neither is it 'displaying the raw data'. On the contrary - > it is corrupting the data with frequencies which shouldn't be there AND > confusing the human eye (for which, presumably, we are doing the drawing). > The raw numbers are useful - indeed essential - for the computer but > fundamentally cannot just 'be displayed' to the human eye as an image. Our > sampling rationale is based on sine-wave frequencies and therefore, as Alvy > Ray Smith said, sinusoidal mapping is the truest (not the most aesthetic, > though this is also true) way of displaying the data. It doesn't add any > spurious higher harmonics, it presents the data as accurately as our > sampling permits. Drawing little boxes may be easier, but it is just as > much mapping the measured samples to a displayed image - the difference is > that this method is both inaccurate and un-aesthetic. > > > > If your CD player spat out square waves to the speakers, you'd take it > back to the shop pretty promptly! > > > > > Guy > > > > -----Original Message----- > > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Mark Cannell > > Sent: Monday, 16 April 2012 5:27 PM > > To: [hidden email] > > Subject: Re: A pixel is not a little square > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > I think I see the problem, the spurious frequencies arise from your > thinking the _data_ is a continuous function and treating it as such (by > "drawing a line ..."), but it is not, it is discrete and can be faithfully > represented by a _discrete_ Fourier transform (which folds at Fs/2). The > hiighest frequency in the DFT is Fs, but we know we shouldn't look at that > right? There are _no_ 'higher harmonics' present in the data, only in > ones 'artistic' interpretation for display purposes. > > > > If it looks jagged, that is because in reality sampled data really is! > The problem really arises because you do not know how to fill in the space > between data samples. You can interpolate (or not). If you interpolate you > are making a statement about the model underlying the data and have just > carried out a fitting exercise. Fitting is NOT raw data presentation. If > you just plot data values you make no assumption about what should join the > data, no model has been fit to the data. Every scientist should know the > difference between a histogram and a continuous distribution and not be > fooled by the vertical lines at the histogram boundaries (which is what you > show in a pixel image). > > > > The choice is yours, in one case you faithfully show unadulterated > sampled data (the histogram looks less 'pretty' than a curve) or you fit a > model and interpolate. The trouble with the latter is that the model is > probably wrong and you hide the defects in the data (e.g. camera pixel > size) from the keen eyed reviewer... Of course if the data points are > really close together, the myopic reviewer can't see defects in you data > :-) ! From Guy's reasoning, it would be impossible to represent any > digitally sampled data because you are always pixelating a continuous > function (all pictures get mad up of little squares -the printer dumps > blobs of ink etc). So, where does the pixelation become acceptable? This is > now aesthetic and has nothing to do with science or mathematics (those with > perfect vision will always see discretization 'artifacts' more easily) . > > > > Cheers Mark > > > > On 16/04/2012, at 3:31 AM, Guy Cox wrote: > > > >> ***** > >> To join, leave or search the confocal microscopy listserv, go to: > >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >> ***** > >> > >> OK, having slept on it, I now feel that just maybe I can explain what > this is all about. If only the list would let us include pictures it would > be much easier! > >> > >> Let's assume we have a digital image, from any source, consisting of > pixels with a spacing s. The smallest spacing we can resolve in this image > is 2s, and this will correspond, in frequency space, with a frequency f. f > represents the bandpass limit of this system, no higher frequencies can be > passed. Now imagine we have a row of pixels containing the following > values: > >> > >> 255 0 255 0 255 0 255 0 255 > >> > >> If we represent these pixels by little squares, we'll have something > like a chessboard. Taking a line along this chessboard will give us a > square wave. Now this square wave cannot be represented within the > bandpass limit of the system, defined by the frequency f. To represent a > square wave we need an infinite series of sine waves f + 3f + 5f +7f ..... > To get even a crude approximation to a square wave we need f + 3f - that > is a frequency three times higher than the image can contain. > >> > >> In other words, we've introduced a whole series of spurious frequencies > into our image that not only were not there to start with, they could not > possibly have been there. Does this matter? After all, we know they > can't be real. It does matter, because we are talking about a visual > representation of our data - that's why we drew the little boxes in the > first place. Our eyes are very sensitive to edges* and the edges will take > over if we let these frequencies come within the bandwidth of our eyes. > We will find it very hard to actually see the finest detail in our picture > (defined by 2s, remember) because if we enlarge it enough to see this > easily we'll also get the edges created by these spurious frequencies. In > everyday terms, the pixellation takes over from the picture. > >> > >> Note that in all this discussion I have not mentioned microscopes, > cameras or anything - we are just talking about a digital image from any > source. It applies to confocal, widefield, and electron microscopes, > telescopes, X-ray images and your holiday snaps. Coming back to the > microscopic world, if we oversample to the point where r, our minimum > resolved distance, is substantially greater than 2s, we may not need to > enlarge to the point where we see the spurious frequencies. This is > probably why some contributors to this discussion have advocated > considerable levels of oversampling (though they probably didn't realise > this, they just knew they got good pictures that way). But oversampling in > fluorescence can be very hard on our specimens. > >> > >> "But I'm using a CCD detector so my image is made up of little > squares". Yes, you can produce a 'coloured in' picture of your detector > that way. I'm assuming the image is actually what you want to see, though, > not the detector. > >> > >> *Amusingly, the human eye does the same thing to emphasize edges as > computer image processing does - it makes the dark side of the edge darker > than it is and the light side lighter. > >> > >> > Guy > >> > >> PS. This has doubtless confirmed my reputation among some people as an > arrogant bastard. They are probably right, but at least I'm an arrogant > bastard who tries to help. It's taken me two hours to write this. > |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Christophe, As I said in my original message, if you want to produce an image of your sensor, with the squares coloured in, that is indeed the case - no argument. But that is not an image of your sample, it's an image of your detector. I was (and am) writing about the digital image in general, as I thought I made clear, regardless of the detection technology. Your CCD image does NOT tell you that intensity X exists across the whole of one detector element, and then suddenly intensity Y exists across the whole of the next detector element. As you said yourself "There is no information about where did the photons arrive on the square detector portion corresponding to a pixel". So we are still dealing with samples and have to handle the problem of relating these samples to the real world. Guy -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christophe Leterrier Sent: Monday, 16 April 2012 10:52 PM To: [hidden email] Subject: Re: A pixel is not a little square ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Guy, But in the CCD case the sampling is integrated over the whole pixel area of the sensor (assuming pixel coverage of 100% which is the case for EMCCDs for exemple). There is no information about where did the photons arrive on the square detector portion corresponding to a pixel. The value X you get as intensity for this pixel is the integrated density over the whole detector portion, so why would you take that value as a point-measurement corresponding to the center of this area ? In that CCD case, isn't the mage indeed an histogram with a bin that has (quite literally) a width equal to the pixel size, not a sample at mathematical points corresponding to the centers of the pixels ? Christophe On Mon, Apr 16, 2012 at 14:36, Guy Cox <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Mark, > > You are continuing to confuse the samples with the > representation of those samples. > > Let's imagine we have a series of data points: 100 90 80 70 60 > 50 40 > > We are mapping these on an image where each is separated by a defined > distance. So we need to fill in this distance. > > You are saying that the 'correct' representation is: > 100 100 100 100 100 90 90 90 90 90 80 80 80 80 80 > 70 70 70 70 60 60 60 60 60 50 50 50 50 50 40 > > I am saying that this is a wildly implausible and totally unjustified > interpretation, and the best representation we can derive from the data is: > 100 98 96 94 92 90 88 86 84 82 80 78 76 74 72 > 70 68 66 64 62 60 58 56 54 52 50 48 46 44 42 > 40 > > EITHER way we are interpolating the sampled data - we have no option - so > let's just get over this. Your proposed representation includes detail > that we could not possibly detect, mine does not. Remember, these are > SAMPLES. Neither representation changes our recorded data. End of > story, IMHO. > > How did we get into this mess? Why does everyone then 'do it wrong'? > Well, actually, everyone doesn't. Scanning probe microscopes always > remap > - because by the time they appeared the computing power to do it was > available. When confocal microscopes first became widely available, > in 1987, the data they produced completely overwhelmed available computing > power (believe me, I was there and writing software). So we got used to > the 'quick and dirty' approach. Consumer digital cameras do a sort of > remap because the Bayer mosaic requires it, but modern sensors so far > exceed the resolution of the camera optics that we never get to see > any spurious frequencies anyway. Computer games consoles always > remap. So do X-ray and EM tomography systems. > > Guy (arrogant bastard) > > > > ----Original Message----- > From: Confocal Microscopy List > [mailto:[hidden email]] > On Behalf Of Mark Cannell > Sent: Monday, 16 April 2012 7:37 PM > To: [hidden email] > Subject: Re: A pixel is not a little square > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Sorry Guy, I still think you don't see the point I'm trying to make. > The camera actually says "The mean signal from x to x+dx is ..." > (where dx is the sensor pixel size). It does NOT say the signal at x > is 'K' and that is where I think the confusion lies. The camera output > is a 2D 'histogram' and showing little boxes with the same intensity > is (I say again) a perfectly accurate representation of the data ( > i.e. F(x) for x -> x + dx = K). With respect, it is not, as you say > inaccurate -even if it is unaesthetic. If you fit a sinusoid you have > just carried out a fitting exercise... That is not a "more accurate" > presentation of the data despite what your Smith says (even if it may > be a more accurate representation of the object which has been > discretized). One should not loose sight of the fact that you have made some (possibly large) assumptions in the fitting process. > > Put mathematically, if you smooth out the displayed pixel edges you > extend the actual sampling frequency (note how you are putting new > unrecorded samples between recorded data values -which is what > drawing a line between points actually does) -you are adding > information to the data that was NOT present in the RAW data. It may > be that your additional information is correct and adds value (e.g. > the band limit of the microscope is...) but one should not loose sight > of distinction between the addition of data/information by the > experimenter (which may or may not be wrong) and that reported by the > instrument (the closest to truth the experimenter can get). > > At the risk of boring some readers on this list, let me emphasize my > point > : The camera actually says "The mean signal from x to x+dx is ..." > (where dx is the sensor pixel size). It does NOT say the signal at x > is 'K' . This can be portrayed as a square with constant color and I > can think of no other truer portrayal of the measured data. Hopefully > dx is less than the resolution of the viewer at final display > resolution but if it is not, then the only choice (IMHO) is between > aesthetics (or some other goal) and truthfully displaying the recorded data -there is no middle ground. > > Cheers Mark > > PS My CD player can't output square waves because the detector etc. > has a rather finite bandwidth... Even if it could, my ears are too > many dB down at 44 kHz to sample it correctly and hear the artifacts > introduced by digital sampling ... :-) > > On 16/04/2012, at 9:18 AM, Guy Cox wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > " There are _no_ 'higher harmonics' present in the data, only in > > ones > 'artistic' interpretation for display purposes." That is exactly what > I said! > > > > I also never said that the data is a continuous function, I said it > > is a > series of discrete samples of a continuous function. So when you > choose to display it you have to do something. Drawing little boxes > is NOT 'doing nothing' and neither is it 'displaying the raw data'. > On the contrary - it is corrupting the data with frequencies which > shouldn't be there AND confusing the human eye (for which, presumably, we are doing the drawing). > The raw numbers are useful - indeed essential - for the computer but > fundamentally cannot just 'be displayed' to the human eye as an image. > Our sampling rationale is based on sine-wave frequencies and > therefore, as Alvy Ray Smith said, sinusoidal mapping is the truest > (not the most aesthetic, though this is also true) way of displaying > the data. It doesn't add any spurious higher harmonics, it presents > the data as accurately as our sampling permits. Drawing little boxes > may be easier, but it is just as much mapping the measured samples to > a displayed image - the difference is that this method is both inaccurate and un-aesthetic. > > > > If your CD player spat out square waves to the speakers, you'd take > > it > back to the shop pretty promptly! > > > > > Guy > > > > -----Original Message----- > > From: Confocal Microscopy List > > [mailto:[hidden email]] > On Behalf Of Mark Cannell > > Sent: Monday, 16 April 2012 5:27 PM > > To: [hidden email] > > Subject: Re: A pixel is not a little square > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > I think I see the problem, the spurious frequencies arise from your > thinking the _data_ is a continuous function and treating it as such > (by "drawing a line ..."), but it is not, it is discrete and can be > faithfully represented by a _discrete_ Fourier transform (which folds > at Fs/2). The hiighest frequency in the DFT is Fs, but we know we > shouldn't look at that right? There are _no_ 'higher harmonics' > present in the data, only in ones 'artistic' interpretation for display purposes. > > > > If it looks jagged, that is because in reality sampled data really is! > The problem really arises because you do not know how to fill in the > space between data samples. You can interpolate (or not). If you > interpolate you are making a statement about the model underlying the > data and have just carried out a fitting exercise. Fitting is NOT raw > data presentation. If you just plot data values you make no assumption > about what should join the data, no model has been fit to the data. > Every scientist should know the difference between a histogram and a > continuous distribution and not be fooled by the vertical lines at the > histogram boundaries (which is what you show in a pixel image). > > > > The choice is yours, in one case you faithfully show unadulterated > sampled data (the histogram looks less 'pretty' than a curve) or you > fit a model and interpolate. The trouble with the latter is that the > model is probably wrong and you hide the defects in the data (e.g. > camera pixel > size) from the keen eyed reviewer... Of course if the data points are > really close together, the myopic reviewer can't see defects in you > data > :-) ! From Guy's reasoning, it would be impossible to represent any > digitally sampled data because you are always pixelating a continuous > function (all pictures get mad up of little squares -the printer dumps > blobs of ink etc). So, where does the pixelation become acceptable? > This is now aesthetic and has nothing to do with science or > mathematics (those with perfect vision will always see discretization 'artifacts' more easily) . > > > > Cheers Mark > > > > On 16/04/2012, at 3:31 AM, Guy Cox wrote: > > > >> ***** > >> To join, leave or search the confocal microscopy listserv, go to: > >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >> ***** > >> > >> OK, having slept on it, I now feel that just maybe I can explain > >> what > this is all about. If only the list would let us include pictures it > would be much easier! > >> > >> Let's assume we have a digital image, from any source, consisting > >> of > pixels with a spacing s. The smallest spacing we can resolve in this > image is 2s, and this will correspond, in frequency space, with a > frequency f. f represents the bandpass limit of this system, no > higher frequencies can be passed. Now imagine we have a row of pixels > containing the following > values: > >> > >> 255 0 255 0 255 0 255 0 255 > >> > >> If we represent these pixels by little squares, we'll have > >> something > like a chessboard. Taking a line along this chessboard will give us a > square wave. Now this square wave cannot be represented within the > bandpass limit of the system, defined by the frequency f. To > represent a square wave we need an infinite series of sine waves f + 3f + 5f +7f ..... > To get even a crude approximation to a square wave we need f + 3f - > that is a frequency three times higher than the image can contain. > >> > >> In other words, we've introduced a whole series of spurious > >> frequencies > into our image that not only were not there to start with, they could not > possibly have been there. Does this matter? After all, we know they > can't be real. It does matter, because we are talking about a visual > representation of our data - that's why we drew the little boxes in > the first place. Our eyes are very sensitive to edges* and the edges > will take over if we let these frequencies come within the bandwidth of our eyes. > We will find it very hard to actually see the finest detail in our > picture (defined by 2s, remember) because if we enlarge it enough to > see this easily we'll also get the edges created by these spurious > frequencies. In everyday terms, the pixellation takes over from the picture. > >> > >> Note that in all this discussion I have not mentioned microscopes, > cameras or anything - we are just talking about a digital image from > any source. It applies to confocal, widefield, and electron > microscopes, telescopes, X-ray images and your holiday snaps. Coming > back to the microscopic world, if we oversample to the point where r, > our minimum resolved distance, is substantially greater than 2s, we > may not need to enlarge to the point where we see the spurious > frequencies. This is probably why some contributors to this > discussion have advocated considerable levels of oversampling (though > they probably didn't realise this, they just knew they got good > pictures that way). But oversampling in fluorescence can be very hard on our specimens. > >> > >> "But I'm using a CCD detector so my image is made up of little > squares". Yes, you can produce a 'coloured in' picture of your > detector that way. I'm assuming the image is actually what you want > to see, though, not the detector. > >> > >> *Amusingly, the human eye does the same thing to emphasize edges as > computer image processing does - it makes the dark side of the edge > darker than it is and the light side lighter. > >> > >> > Guy > >> > >> PS. This has doubtless confirmed my reputation among some people > >> as an > arrogant bastard. They are probably right, but at least I'm an > arrogant bastard who tries to help. It's taken me two hours to write this. > |
|
Tim Holmes |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I am unclear on what is being "figure out". But, the signal processing theory for the type of sampling that CCD elements do is well established in old textbooks on communication theory. The old one on my shelf is by Taub and Schilling entitled Principles of Communication Systems, McGraw-Hill 1971 (but I took the course in 1978 - !!! ;o)). Page 167 gives a nice simple diagram. This type of sampling is called "natural sampling" according to the communication theorists of that time. At least according to ideal theory, the signal can be recovered by theoretical techniques in the book so long as Nyquist is satisfied and there's no noise (which of course is not the case). But, at least we know that, in the ideal case, the information is preserved. I don't know if that's what you are trying to figure, or maybe I'm just making it more confusing ;o) -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Guy Cox Sent: Monday, April 16, 2012 8:16 AM To: [hidden email] Subject: Re: A pixel is not a little square ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Christophe, As I said in my original message, if you want to produce an image of your sensor, with the squares coloured in, that is indeed the case - no argument. But that is not an image of your sample, it's an image of your detector. I was (and am) writing about the digital image in general, as I thought I made clear, regardless of the detection technology. Your CCD image does NOT tell you that intensity X exists across the whole of one detector element, and then suddenly intensity Y exists across the whole of the next detector element. As you said yourself "There is no information about where did the photons arrive on the square detector portion corresponding to a pixel". So we are still dealing with samples and have to handle the problem of relating these samples to the real world. Guy -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christophe Leterrier Sent: Monday, 16 April 2012 10:52 PM To: [hidden email] Subject: Re: A pixel is not a little square ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Guy, But in the CCD case the sampling is integrated over the whole pixel area of the sensor (assuming pixel coverage of 100% which is the case for EMCCDs for exemple). There is no information about where did the photons arrive on the square detector portion corresponding to a pixel. The value X you get as intensity for this pixel is the integrated density over the whole detector portion, so why would you take that value as a point-measurement corresponding to the center of this area ? In that CCD case, isn't the mage indeed an histogram with a bin that has (quite literally) a width equal to the pixel size, not a sample at mathematical points corresponding to the centers of the pixels ? Christophe On Mon, Apr 16, 2012 at 14:36, Guy Cox <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Mark, > > You are continuing to confuse the samples with the > representation of those samples. > > Let's imagine we have a series of data points: 100 90 80 70 60 > 50 40 > > We are mapping these on an image where each is separated by a defined > distance. So we need to fill in this distance. > > You are saying that the 'correct' representation is: > 100 100 100 100 100 90 90 90 90 90 80 80 80 80 80 > 70 70 70 70 60 60 60 60 60 50 50 50 50 50 > > I am saying that this is a wildly implausible and totally unjustified > interpretation, and the best representation we can derive from the data is: > 100 98 96 94 92 90 88 86 84 82 80 78 76 74 72 > 70 68 66 64 62 60 58 56 54 52 50 48 46 44 42 > 40 > > EITHER way we are interpolating the sampled data - we have no option - so > let's just get over this. Your proposed representation includes detail > that we could not possibly detect, mine does not. Remember, these are > SAMPLES. Neither representation changes our recorded data. End of > story, IMHO. > > How did we get into this mess? Why does everyone then 'do it wrong'? > Well, actually, everyone doesn't. Scanning probe microscopes always > remap > - because by the time they appeared the computing power to do it was > available. When confocal microscopes first became widely available, > in 1987, the data they produced completely overwhelmed available computing > power (believe me, I was there and writing software). So we got used to > the 'quick and dirty' approach. Consumer digital cameras do a sort of > remap because the Bayer mosaic requires it, but modern sensors so far > exceed the resolution of the camera optics that we never get to see > any spurious frequencies anyway. Computer games consoles always > remap. So do X-ray and EM tomography systems. > > Guy (arrogant bastard) > > > > ----Original Message----- > From: Confocal Microscopy List > [mailto:[hidden email]] > On Behalf Of Mark Cannell > Sent: Monday, 16 April 2012 7:37 PM > To: [hidden email] > Subject: Re: A pixel is not a little square > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Sorry Guy, I still think you don't see the point I'm trying to make. > The camera actually says "The mean signal from x to x+dx is ..." > (where dx is the sensor pixel size). It does NOT say the signal at x > is 'K' and that is where I think the confusion lies. The camera output > is a 2D 'histogram' and showing little boxes with the same intensity > is (I say again) a perfectly accurate representation of the data ( > i.e. F(x) for x -> x + dx = K). With respect, it is not, as you say > inaccurate -even if it is unaesthetic. If you fit a sinusoid you have > just carried out a fitting exercise... That is not a "more accurate" > presentation of the data despite what your Smith says (even if it may > be a more accurate representation of the object which has been > discretized). One should not loose sight of the fact that you have made > > Put mathematically, if you smooth out the displayed pixel edges you > extend the actual sampling frequency (note how you are putting new > unrecorded samples between recorded data values -which is what > drawing a line between points actually does) -you are adding > information to the data that was NOT present in the RAW data. It may > be that your additional information is correct and adds value (e.g. > the band limit of the microscope is...) but one should not loose sight > of distinction between the addition of data/information by the > experimenter (which may or may not be wrong) and that reported by the > instrument (the closest to truth the experimenter can get). > > At the risk of boring some readers on this list, let me emphasize my > point > : The camera actually says "The mean signal from x to x+dx is ..." > (where dx is the sensor pixel size). It does NOT say the signal at x > is 'K' . This can be portrayed as a square with constant color and I > can think of no other truer portrayal of the measured data. Hopefully > dx is less than the resolution of the viewer at final display > resolution but if it is not, then the only choice (IMHO) is between > aesthetics (or some other goal) and truthfully displaying the recorded > > Cheers Mark > > PS My CD player can't output square waves because the detector etc. > has a rather finite bandwidth... Even if it could, my ears are too > many dB down at 44 kHz to sample it correctly and hear the artifacts > introduced by digital sampling ... :-) > > On 16/04/2012, at 9:18 AM, Guy Cox wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > " There are _no_ 'higher harmonics' present in the data, only in > > ones > 'artistic' interpretation for display purposes." That is exactly what > I said! > > > > I also never said that the data is a continuous function, I said it > > is a > series of discrete samples of a continuous function. So when you > choose to display it you have to do something. Drawing little boxes > is NOT 'doing nothing' and neither is it 'displaying the raw data'. > On the contrary - it is corrupting the data with frequencies which > shouldn't be there AND confusing the human eye (for which, presumably, we > The raw numbers are useful - indeed essential - for the computer but > fundamentally cannot just 'be displayed' to the human eye as an image. > Our sampling rationale is based on sine-wave frequencies and > therefore, as Alvy Ray Smith said, sinusoidal mapping is the truest > (not the most aesthetic, though this is also true) way of displaying > the data. It doesn't add any spurious higher harmonics, it presents > the data as accurately as our sampling permits. Drawing little boxes > may be easier, but it is just as much mapping the measured samples to > a displayed image - the difference is that this method is both inaccurate and un-aesthetic. > > > > If your CD player spat out square waves to the speakers, you'd take > > it > back to the shop pretty promptly! > > > > > Guy > > > > -----Original Message----- > > From: Confocal Microscopy List > > [mailto:[hidden email]] > On Behalf Of Mark Cannell > > Sent: Monday, 16 April 2012 5:27 PM > > To: [hidden email] > > Subject: Re: A pixel is not a little square > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > I think I see the problem, the spurious frequencies arise from your > thinking the _data_ is a continuous function and treating it as such > (by "drawing a line ..."), but it is not, it is discrete and can be > faithfully represented by a _discrete_ Fourier transform (which folds > at Fs/2). The hiighest frequency in the DFT is Fs, but we know we > shouldn't look at that right? There are _no_ 'higher harmonics' > present in the data, only in ones 'artistic' interpretation for display > > > > If it looks jagged, that is because in reality sampled data really is! > The problem really arises because you do not know how to fill in the > space between data samples. You can interpolate (or not). If you > interpolate you are making a statement about the model underlying the > data and have just carried out a fitting exercise. Fitting is NOT raw > data presentation. If you just plot data values you make no assumption > about what should join the data, no model has been fit to the data. > Every scientist should know the difference between a histogram and a > continuous distribution and not be fooled by the vertical lines at the > histogram boundaries (which is what you show in a pixel image). > > > > The choice is yours, in one case you faithfully show unadulterated > sampled data (the histogram looks less 'pretty' than a curve) or you > fit a model and interpolate. The trouble with the latter is that the > model is probably wrong and you hide the defects in the data (e.g. > camera pixel > size) from the keen eyed reviewer... Of course if the data points are > really close together, the myopic reviewer can't see defects in you > data > :-) ! From Guy's reasoning, it would be impossible to represent any > digitally sampled data because you are always pixelating a continuous > function (all pictures get mad up of little squares -the printer dumps > blobs of ink etc). So, where does the pixelation become acceptable? > This is now aesthetic and has nothing to do with science or > mathematics (those with perfect vision will always see discretization > > > > Cheers Mark > > > > On 16/04/2012, at 3:31 AM, Guy Cox wrote: > > > >> ***** > >> To join, leave or search the confocal microscopy listserv, go to: > >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >> ***** > >> > >> OK, having slept on it, I now feel that just maybe I can explain > >> what > this is all about. If only the list would let us include pictures it > would be much easier! > >> > >> Let's assume we have a digital image, from any source, consisting > >> of > pixels with a spacing s. The smallest spacing we can resolve in this > image is 2s, and this will correspond, in frequency space, with a > frequency f. f represents the bandpass limit of this system, no > higher frequencies can be passed. Now imagine we have a row of pixels > containing the following > values: > >> > >> 255 0 255 0 255 0 255 0 255 > >> > >> If we represent these pixels by little squares, we'll have > >> something > like a chessboard. Taking a line along this chessboard will give us a > square wave. Now this square wave cannot be represented within the > bandpass limit of the system, defined by the frequency f. To > represent a square wave we need an infinite series of sine waves f + 3f + > To get even a crude approximation to a square wave we need f + 3f - > that is a frequency three times higher than the image can contain. > >> > >> In other words, we've introduced a whole series of spurious > >> frequencies > into our image that not only were not there to start with, they could not > possibly have been there. Does this matter? After all, we know they > can't be real. It does matter, because we are talking about a visual > representation of our data - that's why we drew the little boxes in > the first place. Our eyes are very sensitive to edges* and the edges > will take over if we let these frequencies come within the bandwidth of > We will find it very hard to actually see the finest detail in our > picture (defined by 2s, remember) because if we enlarge it enough to > see this easily we'll also get the edges created by these spurious > frequencies. In everyday terms, the pixellation takes over from the picture. > >> > >> Note that in all this discussion I have not mentioned microscopes, > cameras or anything - we are just talking about a digital image from > any source. It applies to confocal, widefield, and electron > microscopes, telescopes, X-ray images and your holiday snaps. Coming > back to the microscopic world, if we oversample to the point where r, > our minimum resolved distance, is substantially greater than 2s, we > may not need to enlarge to the point where we see the spurious > frequencies. This is probably why some contributors to this > discussion have advocated considerable levels of oversampling (though > they probably didn't realise this, they just knew they got good > pictures that way). But oversampling in fluorescence can be very hard on > >> > >> "But I'm using a CCD detector so my image is made up of little > squares". Yes, you can produce a 'coloured in' picture of your > detector that way. I'm assuming the image is actually what you want > to see, though, not the detector. > >> > >> *Amusingly, the human eye does the same thing to emphasize edges as > computer image processing does - it makes the dark side of the edge > darker than it is and the light side lighter. > >> > >> > Guy > >> > >> PS. This has doubtless confirmed my reputation among some people > >> as an > arrogant bastard. They are probably right, but at least I'm an > arrogant bastard who tries to help. It's taken me two hours to write > ----- No virus found in this message. Checked by AVG - www.avg.com Version: 2012.0.1913 / Virus Database: 2411/4939 - Release Date: 04/15/12 |
|
Mark Cannell |
|
|
In reply to this post by Guy Cox-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** On 16/04/2012, at 1:36 PM, Guy Cox wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Mark, > > You are continuing to confuse the samples with the representation of those samples. > > Let's imagine we have a series of data points: 100 90 80 70 60 50 40 NOOOOOOOO! They are NOT data points! The camera does not report points at all! You are making the common error of assuming the camera provides a DIrac comb filter. It does not. > > We are mapping these on an image where each is separated by a defined distance. So we need to fill in this distance. > > You are saying that the 'correct' representation is: > 100 100 100 100 100 90 90 90 90 90 80 80 80 80 80 70 70 70 70 60 60 60 60 60 50 50 50 50 50 40 Yes that is what it reported, the average from x to x=dx is a constant. 'Nuf said. Cheers Rest deleted for tedium > > I am saying that this is a wildly implausible and totally unjustified interpretation, and the best representation we can derive from the data is: > 100 98 96 94 92 90 88 86 84 82 80 78 76 74 72 70 68 66 64 62 60 58 56 54 52 50 48 46 44 42 40 > > EITHER way we are interpolating the sampled data - we have no option - so let's just get over this. Your proposed representation includes detail that we could not possibly detect, mine does not. Remember, these are SAMPLES. Neither representation changes our recorded data. End of story, IMHO. > > How did we get into this mess? Why does everyone then 'do it wrong'? Well, actually, everyone doesn't. Scanning probe microscopes always remap - because by the time they appeared the computing power to do it was available. When confocal microscopes first became widely available, in 1987, the data they produced completely overwhelmed available computing power (believe me, I was there and writing software). So we got used to the 'quick and dirty' approach. Consumer digital cameras do a sort of remap because the Bayer mosaic requires it, but modern sensors so far exceed the resolution of the camera optics that we never get to see any spurious frequencies anyway. Computer games consoles always remap. So do X-ray and EM tomography systems. > > Guy (arrogant bastard) > > > > ----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mark Cannell > Sent: Monday, 16 April 2012 7:37 PM > To: [hidden email] > Subject: Re: A pixel is not a little square > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Sorry Guy, I still think you don't see the point I'm trying to make. The camera actually says "The mean signal from x to x+dx is ..." (where dx is the sensor pixel size). It does NOT say the signal at x is 'K' and that is where I think the confusion lies. The camera output is a 2D 'histogram' and showing little boxes with the same intensity is (I say again) a perfectly accurate representation of the data ( i.e. F(x) for x -> x + dx = K). With respect, it is not, as you say inaccurate -even if it is unaesthetic. If you fit a sinusoid you have just carried out a fitting exercise... That is not a "more accurate" presentation of the data despite what your Smith says (even if it may be a more accurate representation of the object which has been discretized). One should not loose sight of the fact that you have made some (possibly large) assumptions in the fitting process. > > Put mathematically, if you smooth out the displayed pixel edges you extend the actual sampling frequency (note how you are putting new unrecorded samples between recorded data values -which is what drawing a line between points actually does) -you are adding information to the data that was NOT present in the RAW data. It may be that your additional information is correct and adds value (e.g. the band limit of the microscope is...) but one should not loose sight of distinction between the addition of data/information by the experimenter (which may or may not be wrong) and that reported by the instrument (the closest to truth the experimenter can get). > > At the risk of boring some readers on this list, let me emphasize my point : The camera actually says "The mean signal from x to x+dx is ..." (where dx is the sensor pixel size). It does NOT say the signal at x is 'K' . This can be portrayed as a square with constant color and I can think of no other truer portrayal of the measured data. Hopefully dx is less than the resolution of the viewer at final display resolution but if it is not, then the only choice (IMHO) is between aesthetics (or some other goal) and truthfully displaying the recorded data -there is no middle ground. > > Cheers Mark > > PS My CD player can't output square waves because the detector etc. has a rather finite bandwidth... Even if it could, my ears are too many dB down at 44 kHz to sample it correctly and hear the artifacts introduced by digital sampling ... :-) > > On 16/04/2012, at 9:18 AM, Guy Cox wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> " There are _no_ 'higher harmonics' present in the data, only in ones 'artistic' interpretation for display purposes." That is exactly what I said! >> >> I also never said that the data is a continuous function, I said it is a series of discrete samples of a continuous function. So when you choose to display it you have to do something. Drawing little boxes is NOT 'doing nothing' and neither is it 'displaying the raw data'. On the contrary - it is corrupting the data with frequencies which shouldn't be there AND confusing the human eye (for which, presumably, we are doing the drawing). The raw numbers are useful - indeed essential - for the computer but fundamentally cannot just 'be displayed' to the human eye as an image. Our sampling rationale is based on sine-wave frequencies and therefore, as Alvy Ray Smith said, sinusoidal mapping is the truest (not the most aesthetic, though this is also true) way of displaying the data. It doesn't add any spurious higher harmonics, it presents the data as accurately as our sampling permits. Drawing little boxes may be easier, but it is just as much mapping the measured samples to a displayed image - the difference is that this method is both inaccurate and un-aesthetic. >> >> If your CD player spat out square waves to the speakers, you'd take it back to the shop pretty promptly! >> >> Guy >> >> -----Original Message----- >> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mark Cannell >> Sent: Monday, 16 April 2012 5:27 PM >> To: [hidden email] >> Subject: Re: A pixel is not a little square >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> I think I see the problem, the spurious frequencies arise from your thinking the _data_ is a continuous function and treating it as such (by "drawing a line ..."), but it is not, it is discrete and can be faithfully represented by a _discrete_ Fourier transform (which folds at Fs/2). The hiighest frequency in the DFT is Fs, but we know we shouldn't look at that right? There are _no_ 'higher harmonics' present in the data, only in ones 'artistic' interpretation for display purposes. >> >> If it looks jagged, that is because in reality sampled data really is! The problem really arises because you do not know how to fill in the space between data samples. You can interpolate (or not). If you interpolate you are making a statement about the model underlying the data and have just carried out a fitting exercise. Fitting is NOT raw data presentation. If you just plot data values you make no assumption about what should join the data, no model has been fit to the data. Every scientist should know the difference between a histogram and a continuous distribution and not be fooled by the vertical lines at the histogram boundaries (which is what you show in a pixel image). >> >> The choice is yours, in one case you faithfully show unadulterated sampled data (the histogram looks less 'pretty' than a curve) or you fit a model and interpolate. The trouble with the latter is that the model is probably wrong and you hide the defects in the data (e.g. camera pixel size) from the keen eyed reviewer... Of course if the data points are really close together, the myopic reviewer can't see defects in you data :-) ! From Guy's reasoning, it would be impossible to represent any digitally sampled data because you are always pixelating a continuous function (all pictures get mad up of little squares -the printer dumps blobs of ink etc). So, where does the pixelation become acceptable? This is now aesthetic and has nothing to do with science or mathematics (those with perfect vision will always see discretization 'artifacts' more easily) . >> >> Cheers Mark >> >> On 16/04/2012, at 3:31 AM, Guy Cox wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> OK, having slept on it, I now feel that just maybe I can explain what this is all about. If only the list would let us include pictures it would be much easier! >>> >>> Let's assume we have a digital image, from any source, consisting of pixels with a spacing s. The smallest spacing we can resolve in this image is 2s, and this will correspond, in frequency space, with a frequency f. f represents the bandpass limit of this system, no higher frequencies can be passed. Now imagine we have a row of pixels containing the following values: >>> >>> 255 0 255 0 255 0 255 0 255 >>> >>> If we represent these pixels by little squares, we'll have something like a chessboard. Taking a line along this chessboard will give us a square wave. Now this square wave cannot be represented within the bandpass limit of the system, defined by the frequency f. To represent a square wave we need an infinite series of sine waves f + 3f + 5f +7f ..... To get even a crude approximation to a square wave we need f + 3f - that is a frequency three times higher than the image can contain. >>> >>> In other words, we've introduced a whole series of spurious frequencies into our image that not only were not there to start with, they could not possibly have been there. Does this matter? After all, we know they can't be real. It does matter, because we are talking about a visual representation of our data - that's why we drew the little boxes in the first place. Our eyes are very sensitive to edges* and the edges will take over if we let these frequencies come within the bandwidth of our eyes. We will find it very hard to actually see the finest detail in our picture (defined by 2s, remember) because if we enlarge it enough to see this easily we'll also get the edges created by these spurious frequencies. In everyday terms, the pixellation takes over from the picture. >>> >>> Note that in all this discussion I have not mentioned microscopes, cameras or anything - we are just talking about a digital image from any source. It applies to confocal, widefield, and electron microscopes, telescopes, X-ray images and your holiday snaps. Coming back to the microscopic world, if we oversample to the point where r, our minimum resolved distance, is substantially greater than 2s, we may not need to enlarge to the point where we see the spurious frequencies. This is probably why some contributors to this discussion have advocated considerable levels of oversampling (though they probably didn't realise this, they just knew they got good pictures that way). But oversampling in fluorescence can be very hard on our specimens. >>> >>> "But I'm using a CCD detector so my image is made up of little squares". Yes, you can produce a 'coloured in' picture of your detector that way. I'm assuming the image is actually what you want to see, though, not the detector. >>> >>> *Amusingly, the human eye does the same thing to emphasize edges as computer image processing does - it makes the dark side of the edge darker than it is and the light side lighter. >>> >>> Guy >>> >>> PS. This has doubtless confirmed my reputation among some people as an arrogant bastard. They are probably right, but at least I'm an arrogant bastard who tries to help. It's taken me two hours to write this. |
|
George McNamara |
|
|
In reply to this post by Guy Cox-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Guy, I feel better knowing you are an even slower typist (or explainer?) than I am. Thanks, George On 4/15/2012 10:31 PM, Guy Cox wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > OK, having slept on it, I now feel that just maybe I can explain what this is all about. If only the list would let us include pictures it would be much easier! > > Let's assume we have a digital image, from any source, consisting of pixels with a spacing s. The smallest spacing we can resolve in this image is 2s, and this will correspond, in frequency space, with a frequency f. f represents the bandpass limit of this system, no higher frequencies can be passed. Now imagine we have a row of pixels containing the following values: > > 255 0 255 0 255 0 255 0 255 > > If we represent these pixels by little squares, we'll have something like a chessboard. Taking a line along this chessboard will give us a square wave. Now this square wave cannot be represented within the bandpass limit of the system, defined by the frequency f. To represent a square wave we need an infinite series of sine waves f + 3f + 5f +7f ..... To get even a crude approximation to a square wave we need f + 3f - that is a frequency three times higher than the image can contain. > > In other words, we've introduced a whole series of spurious frequencies into our image that not only were not there to start with, they could not possibly have been there. Does this matter? After all, we know they can't be real. It does matter, because we are talking about a visual representation of our data - that's why we drew the little boxes in the first place. Our eyes are very sensitive to edges* and the edges will take over if we let these frequencies come within the bandwidth of our eyes. We will find it very hard to actually see the finest detail in our picture (defined by 2s, remember) because if we enlarge it enough to see this easily we'll also get the edges created by these spurious frequencies. In everyday terms, the pixellation takes over from the picture. > > Note that in all this discussion I have not mentioned microscopes, cameras or anything - we are just talking about a digital image from any source. It applies to confocal, widefield, and electron microscopes, telescopes, X-ray images and your holiday snaps. Coming back to the microscopic world, if we oversample to the point where r, our minimum resolved distance, is substantially greater than 2s, we may not need to enlarge to the point where we see the spurious frequencies. This is probably why some contributors to this discussion have advocated considerable levels of oversampling (though they probably didn't realise this, they just knew they got good pictures that way). But oversampling in fluorescence can be very hard on our specimens. > > "But I'm using a CCD detector so my image is made up of little squares". Yes, you can produce a 'coloured in' picture of your detector that way. I'm assuming the image is actually what you want to see, though, not the detector. > > *Amusingly, the human eye does the same thing to emphasize edges as computer image processing does - it makes the dark side of the edge darker than it is and the light side lighter. > > Guy > > PS. This has doubtless confirmed my reputation among some people as an arrogant bastard. They are probably right, but at least I'm an arrogant bastard who tries to help. It's taken me two hours to write this. > > |
|
|
In reply to this post by Mark Cannell
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Mark, We are totally at cross purposes here. Maybe you are cranky with me for using the word bollocks (which I thought was part of robust Antipodean discourse). If so, I apologise unreservedly. I was not making any assumptions about the response of a CCD camera, I was not concerned with any data source whatsoever. Actually, I would really like to know the response of a CCD camera - I'd imagine that pixels on an interline transfer camera (which is what most of us use most of the time) are fairly strongly centre-weighted. Does anyone know any publications on this? As I posted (twice, I think) if we want to present a map of what our CCD has recorded what you suggest is totally correct, but this is an image of our detector, not our specimen. What I was posting about was an image in the abstract - an array of numbers - samples - representing a 2D object that exists in the real world. And I did spell this out. Many years ago I came to realize that what we were doing was not correct (partly because I was a bit involved in scanning probe microscopy) and Colin Sheppard and I had many discussions on this, back when he was in Sydney. Then, last year, I came across Alvy Ray Smith's paper and realised that what I wanted to say had already been said! So I wanted to bring it to the attention of the microscopy community. Alvy's paper is a bit mathematical, so I tried to make it less so. Maybe I wasn't totally successful in this, but actually I first learnt about harmonic series (and how to make a square wave from sine waves) in music class, not maths! Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Guy Cox, MA, DPhil(Oxon), Honorary Associate, Australian Centre for Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW 2006 Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mark Cannell Sent: Tuesday, 17 April 2012 12:29 AM To: [hidden email] Subject: Re: A pixel is not a little square ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** On 16/04/2012, at 1:36 PM, Guy Cox wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Mark, > > You are continuing to confuse the samples with the representation of those samples. > > Let's imagine we have a series of data points: 100 90 80 70 60 50 40 NOOOOOOOO! They are NOT data points! The camera does not report points at all! You are making the common error of assuming the camera provides a DIrac comb filter. It does not. > > We are mapping these on an image where each is separated by a defined distance. So we need to fill in this distance. > > You are saying that the 'correct' representation is: > 100 100 100 100 100 90 90 90 90 90 80 80 80 80 80 70 70 70 70 60 60 60 60 60 50 50 50 50 50 40 Yes that is what it reported, the average from x to x=dx is a constant. 'Nuf said. Cheers Rest deleted for tedium > > I am saying that this is a wildly implausible and totally unjustified interpretation, and the best representation we can derive from the data is: > 100 98 96 94 92 90 88 86 84 82 80 78 76 74 72 70 68 66 64 62 60 58 56 54 52 50 48 46 44 42 40 > > EITHER way we are interpolating the sampled data - we have no option - so let's just get over this. Your proposed representation includes detail that we could not possibly detect, mine does not. Remember, these are SAMPLES. Neither representation changes our recorded data. End of story, IMHO. > > How did we get into this mess? Why does everyone then 'do it wrong'? Well, actually, everyone doesn't. Scanning probe microscopes always remap - because by the time they appeared the computing power to do it was available. When confocal microscopes first became widely available, in 1987, the data they produced completely overwhelmed available computing power (believe me, I was there and writing software). So we got used to the 'quick and dirty' approach. Consumer digital cameras do a sort of remap because the Bayer mosaic requires it, but modern sensors so far exceed the resolution of the camera optics that we never get to see any spurious frequencies anyway. Computer games consoles always remap. So do X-ray and EM tomography systems. > > Guy (arrogant bastard) > > > > ----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mark Cannell > Sent: Monday, 16 April 2012 7:37 PM > To: [hidden email] > Subject: Re: A pixel is not a little square > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Sorry Guy, I still think you don't see the point I'm trying to make. The camera actually says "The mean signal from x to x+dx is ..." (where dx is the sensor pixel size). It does NOT say the signal at x is 'K' and that is where I think the confusion lies. The camera output is a 2D 'histogram' and showing little boxes with the same intensity is (I say again) a perfectly accurate representation of the data ( i.e. F(x) for x -> x + dx = K). With respect, it is not, as you say inaccurate -even if it is unaesthetic. If you fit a sinusoid you have just carried out a fitting exercise... That is not a "more accurate" presentation of the data despite what your Smith says (even if it may be a more accurate representation of the object which has been discretized). One should not loose sight of the fact that you have made some (possibly large) assumptions in the fitting process. > > Put mathematically, if you smooth out the displayed pixel edges you extend the actual sampling frequency (note how you are putting new unrecorded samples between recorded data values -which is what drawing a line between points actually does) -you are adding information to the data that was NOT present in the RAW data. It may be that your additional information is correct and adds value (e.g. the band limit of the microscope is...) but one should not loose sight of distinction between the addition of data/information by the experimenter (which may or may not be wrong) and that reported by the instrument (the closest to truth the experimenter can get). > > At the risk of boring some readers on this list, let me emphasize my point : The camera actually says "The mean signal from x to x+dx is ..." (where dx is the sensor pixel size). It does NOT say the signal at x is 'K' . This can be portrayed as a square with constant color and I can think of no other truer portrayal of the measured data. Hopefully dx is less than the resolution of the viewer at final display resolution but if it is not, then the only choice (IMHO) is between aesthetics (or some other goal) and truthfully displaying the recorded data -there is no middle ground. > > Cheers Mark > > PS My CD player can't output square waves because the detector etc. has a rather finite bandwidth... Even if it could, my ears are too many dB down at 44 kHz to sample it correctly and hear the artifacts introduced by digital sampling ... :-) > > On 16/04/2012, at 9:18 AM, Guy Cox wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> " There are _no_ 'higher harmonics' present in the data, only in ones 'artistic' interpretation for display purposes." That is exactly what I said! >> >> I also never said that the data is a continuous function, I said it is a series of discrete samples of a continuous function. So when you choose to display it you have to do something. Drawing little boxes is NOT 'doing nothing' and neither is it 'displaying the raw data'. On the contrary - it is corrupting the data with frequencies which shouldn't be there AND confusing the human eye (for which, presumably, we are doing the drawing). The raw numbers are useful - indeed essential - for the computer but fundamentally cannot just 'be displayed' to the human eye as an image. Our sampling rationale is based on sine-wave frequencies and therefore, as Alvy Ray Smith said, sinusoidal mapping is the truest (not the most aesthetic, though this is also true) way of displaying the data. It doesn't add any spurious higher harmonics, it presents the data as accurately as our sampling permits. Drawing little boxes may be easier, but it is just as much mapping the measured samples to a displayed image - the difference is that this method is both inaccurate and un-aesthetic. >> >> If your CD player spat out square waves to the speakers, you'd take it back to the shop pretty promptly! >> >> Guy >> >> -----Original Message----- >> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mark Cannell >> Sent: Monday, 16 April 2012 5:27 PM >> To: [hidden email] >> Subject: Re: A pixel is not a little square >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> I think I see the problem, the spurious frequencies arise from your thinking the _data_ is a continuous function and treating it as such (by "drawing a line ..."), but it is not, it is discrete and can be faithfully represented by a _discrete_ Fourier transform (which folds at Fs/2). The hiighest frequency in the DFT is Fs, but we know we shouldn't look at that right? There are _no_ 'higher harmonics' present in the data, only in ones 'artistic' interpretation for display purposes. >> >> If it looks jagged, that is because in reality sampled data really is! The problem really arises because you do not know how to fill in the space between data samples. You can interpolate (or not). If you interpolate you are making a statement about the model underlying the data and have just carried out a fitting exercise. Fitting is NOT raw data presentation. If you just plot data values you make no assumption about what should join the data, no model has been fit to the data. Every scientist should know the difference between a histogram and a continuous distribution and not be fooled by the vertical lines at the histogram boundaries (which is what you show in a pixel image). >> >> The choice is yours, in one case you faithfully show unadulterated sampled data (the histogram looks less 'pretty' than a curve) or you fit a model and interpolate. The trouble with the latter is that the model is probably wrong and you hide the defects in the data (e.g. camera pixel size) from the keen eyed reviewer... Of course if the data points are really close together, the myopic reviewer can't see defects in you data :-) ! From Guy's reasoning, it would be impossible to represent any digitally sampled data because you are always pixelating a continuous function (all pictures get mad up of little squares -the printer dumps blobs of ink etc). So, where does the pixelation become acceptable? This is now aesthetic and has nothing to do with science or mathematics (those with perfect vision will always see discretization 'artifacts' more easily) . >> >> Cheers Mark >> >> On 16/04/2012, at 3:31 AM, Guy Cox wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> OK, having slept on it, I now feel that just maybe I can explain what this is all about. If only the list would let us include pictures it would be much easier! >>> >>> Let's assume we have a digital image, from any source, consisting of pixels with a spacing s. The smallest spacing we can resolve in this image is 2s, and this will correspond, in frequency space, with a frequency f. f represents the bandpass limit of this system, no higher frequencies can be passed. Now imagine we have a row of pixels containing the following values: >>> >>> 255 0 255 0 255 0 255 0 255 >>> >>> If we represent these pixels by little squares, we'll have something like a chessboard. Taking a line along this chessboard will give us a square wave. Now this square wave cannot be represented within the bandpass limit of the system, defined by the frequency f. To represent a square wave we need an infinite series of sine waves f + 3f + 5f +7f ..... To get even a crude approximation to a square wave we need f + 3f - that is a frequency three times higher than the image can contain. >>> >>> In other words, we've introduced a whole series of spurious frequencies into our image that not only were not there to start with, they could not possibly have been there. Does this matter? After all, we know they can't be real. It does matter, because we are talking about a visual representation of our data - that's why we drew the little boxes in the first place. Our eyes are very sensitive to edges* and the edges will take over if we let these frequencies come within the bandwidth of our eyes. We will find it very hard to actually see the finest detail in our picture (defined by 2s, remember) because if we enlarge it enough to see this easily we'll also get the edges created by these spurious frequencies. In everyday terms, the pixellation takes over from the picture. >>> >>> Note that in all this discussion I have not mentioned microscopes, cameras or anything - we are just talking about a digital image from any source. It applies to confocal, widefield, and electron microscopes, telescopes, X-ray images and your holiday snaps. Coming back to the microscopic world, if we oversample to the point where r, our minimum resolved distance, is substantially greater than 2s, we may not need to enlarge to the point where we see the spurious frequencies. This is probably why some contributors to this discussion have advocated considerable levels of oversampling (though they probably didn't realise this, they just knew they got good pictures that way). But oversampling in fluorescence can be very hard on our specimens. >>> >>> "But I'm using a CCD detector so my image is made up of little squares". Yes, you can produce a 'coloured in' picture of your detector that way. I'm assuming the image is actually what you want to see, though, not the detector. >>> >>> *Amusingly, the human eye does the same thing to emphasize edges as computer image processing does - it makes the dark side of the edge darker than it is and the light side lighter. >>> >>> Guy >>> >>> PS. This has doubtless confirmed my reputation among some people as an arrogant bastard. They are probably right, but at least I'm an arrogant bastard who tries to help. It's taken me two hours to write this. |
|
Mark Cannell |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Guy I think the essence of my point is that one should always display minimally processed data (if possible). The trouble with interpolation is that it is rarely (never?) correct (e.g. your example shows a change in intensity and/or a phase shift and/or edge artifact) . Even if a camera were a true sampling device (a Dirak comb) the correct reconstruction which allows resampling for display is convolution with a sinc function (sin(x)/x) which cannot be applied to a finite data set (the sine component has no boundaries). A windowed sinc function my be acceptable in view of the finite data length but it is not rigorous and introduces edge effects that can propagate and assumes a true band limited signal ( -what did the noise actually do to the band limit?). A bicubic spline is easy but of uncertain accuracy when you have an instrument that integrates over space (how do modern confocals actually integrate over the pixel dwell time?). Let me be clear, I'm not against interpolation for display purposes (I've used it myself) but I worry that it can easily hide a multitude of scientific 'sins' therefore getting better raw data is generally preferable IMHO. I also worry that by offering strong advocacy for resampling you encourage sloppy sampling (will anyone ever see the sampling error(s) in the final data?). So if I seem to state my position so strongly, it is because I don't care how pretty the image looks, I just want to see uncorrupted data -we are doing science! By the way, for those that don't know, you can easily suppress pixel edge effect in images (if they offend) by squinting at the image... I suggest that if you want to show an image with lots of empty magnification (so it's smooth) near the resolution limit it would be better to increase magnification on the instrument... ( a benefit of this approach may arise from higher mag objectives tending to have higher NA, so they may also help you realize that there were some problems in the low magnification image or, better yet, more detail). Finally, the expanded pixel image is not an 'image of the sensor' -it's what the recorded data really looks like before an attempt at reconstruction (but I've said that before...)! On this last point I guess we'll just have to agree to disagree. Cheers Mark On 17/04/2012, at 11:16 AM, Guy Cox wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Mark, > > We are totally at cross purposes here. Maybe you are cranky with me for using the word bollocks (which I thought was part of robust Antipodean discourse). If so, I apologise unreservedly. > > I was not making any assumptions about the response of a CCD camera, I was not concerned with any data source whatsoever. Actually, I would really like to know the response of a CCD camera - I'd imagine that pixels on an interline transfer camera (which is what most of us use most of the time) are fairly strongly centre-weighted. Does anyone know any publications on this? As I posted (twice, I think) if we want to present a map of what our CCD has recorded what you suggest is totally correct, but this is an image of our detector, not our specimen. > > What I was posting about was an image in the abstract - an array of numbers - samples - representing a 2D object that exists in the real world. And I did spell this out. Many years ago I came to realize that what we were doing was not correct (partly because I was a bit involved in scanning probe microscopy) and Colin Sheppard and I had many discussions on this, back when he was in Sydney. Then, last year, I came across Alvy Ray Smith's paper and realised that what I wanted to say had already been said! So I wanted to bring it to the attention of the microscopy community. > > Alvy's paper is a bit mathematical, so I tried to make it less so. Maybe I wasn't totally successful in this, but actually I first learnt about harmonic series (and how to make a square wave from sine waves) in music class, not maths! > > Guy > > > > > Optical Imaging Techniques in Cell Biology > by Guy Cox CRC Press / Taylor & Francis > http://www.guycox.com/optical.htm > ______________________________________________ > Guy Cox, MA, DPhil(Oxon), Honorary Associate, > Australian Centre for Microscopy & Microanalysis, > Madsen Building F09, University of Sydney, NSW 2006 > > Phone +61 2 9351 3176 Fax +61 2 9351 7682 > Mobile 0413 281 861 > ______________________________________________ > http://www.guycox.net > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mark Cannell > Sent: Tuesday, 17 April 2012 12:29 AM > To: [hidden email] > Subject: Re: A pixel is not a little square > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > On 16/04/2012, at 1:36 PM, Guy Cox wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Mark, >> >> You are continuing to confuse the samples with the representation of those samples. >> >> Let's imagine we have a series of data points: 100 90 80 70 60 50 40 > > > NOOOOOOOO! They are NOT data points! The camera does not report points at all! You are making the common error of assuming the camera provides a DIrac comb filter. It does not. > > >> >> We are mapping these on an image where each is separated by a defined distance. So we need to fill in this distance. >> >> You are saying that the 'correct' representation is: >> 100 100 100 100 100 90 90 90 90 90 80 80 80 80 80 70 70 70 70 60 60 60 60 60 50 50 50 50 50 40 > > Yes that is what it reported, the average from x to x=dx is a constant. > > 'Nuf said. > > Cheers > > > Rest deleted for tedium >> >> I am saying that this is a wildly implausible and totally unjustified interpretation, and the best representation we can derive from the data is: >> 100 98 96 94 92 90 88 86 84 82 80 78 76 74 72 70 68 66 64 62 60 58 56 54 52 50 48 46 44 42 40 >> >> EITHER way we are interpolating the sampled data - we have no option - so let's just get over this. Your proposed representation includes detail that we could not possibly detect, mine does not. Remember, these are SAMPLES. Neither representation changes our recorded data. End of story, IMHO. >> >> How did we get into this mess? Why does everyone then 'do it wrong'? Well, actually, everyone doesn't. Scanning probe microscopes always remap - because by the time they appeared the computing power to do it was available. When confocal microscopes first became widely available, in 1987, the data they produced completely overwhelmed available computing power (believe me, I was there and writing software). So we got used to the 'quick and dirty' approach. Consumer digital cameras do a sort of remap because the Bayer mosaic requires it, but modern sensors so far exceed the resolution of the camera optics that we never get to see any spurious frequencies anyway. Computer games consoles always remap. So do X-ray and EM tomography systems. >> >> Guy (arrogant bastard) >> >> >> >> ----Original Message----- >> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mark Cannell >> Sent: Monday, 16 April 2012 7:37 PM >> To: [hidden email] >> Subject: Re: A pixel is not a little square >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Sorry Guy, I still think you don't see the point I'm trying to make. The camera actually says "The mean signal from x to x+dx is ..." (where dx is the sensor pixel size). It does NOT say the signal at x is 'K' and that is where I think the confusion lies. The camera output is a 2D 'histogram' and showing little boxes with the same intensity is (I say again) a perfectly accurate representation of the data ( i.e. F(x) for x -> x + dx = K). With respect, it is not, as you say inaccurate -even if it is unaesthetic. If you fit a sinusoid you have just carried out a fitting exercise... That is not a "more accurate" presentation of the data despite what your Smith says (even if it may be a more accurate representation of the object which has been discretized). One should not loose sight of the fact that you have made some (possibly large) assumptions in the fitting process. >> >> Put mathematically, if you smooth out the displayed pixel edges you extend the actual sampling frequency (note how you are putting new unrecorded samples between recorded data values -which is what drawing a line between points actually does) -you are adding information to the data that was NOT present in the RAW data. It may be that your additional information is correct and adds value (e.g. the band limit of the microscope is...) but one should not loose sight of distinction between the addition of data/information by the experimenter (which may or may not be wrong) and that reported by the instrument (the closest to truth the experimenter can get). >> >> At the risk of boring some readers on this list, let me emphasize my point : The camera actually says "The mean signal from x to x+dx is ..." (where dx is the sensor pixel size). It does NOT say the signal at x is 'K' . This can be portrayed as a square with constant color and I can think of no other truer portrayal of the measured data. Hopefully dx is less than the resolution of the viewer at final display resolution but if it is not, then the only choice (IMHO) is between aesthetics (or some other goal) and truthfully displaying the recorded data -there is no middle ground. >> >> Cheers Mark >> >> PS My CD player can't output square waves because the detector etc. has a rather finite bandwidth... Even if it could, my ears are too many dB down at 44 kHz to sample it correctly and hear the artifacts introduced by digital sampling ... :-) >> >> On 16/04/2012, at 9:18 AM, Guy Cox wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> " There are _no_ 'higher harmonics' present in the data, only in ones 'artistic' interpretation for display purposes." That is exactly what I said! >>> >>> I also never said that the data is a continuous function, I said it is a series of discrete samples of a continuous function. So when you choose to display it you have to do something. Drawing little boxes is NOT 'doing nothing' and neither is it 'displaying the raw data'. On the contrary - it is corrupting the data with frequencies which shouldn't be there AND confusing the human eye (for which, presumably, we are doing the drawing). The raw numbers are useful - indeed essential - for the computer but fundamentally cannot just 'be displayed' to the human eye as an image. Our sampling rationale is based on sine-wave frequencies and therefore, as Alvy Ray Smith said, sinusoidal mapping is the truest (not the most aesthetic, though this is also true) way of displaying the data. It doesn't add any spurious higher harmonics, it presents the data as accurately as our sampling permits. Drawing little boxes may be easier, but it is just as much mapping the measured samples to a displayed image - the difference is that this method is both inaccurate and un-aesthetic. >>> >>> If your CD player spat out square waves to the speakers, you'd take it back to the shop pretty promptly! >>> >>> Guy >>> >>> -----Original Message----- >>> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mark Cannell >>> Sent: Monday, 16 April 2012 5:27 PM >>> To: [hidden email] >>> Subject: Re: A pixel is not a little square >>> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> I think I see the problem, the spurious frequencies arise from your thinking the _data_ is a continuous function and treating it as such (by "drawing a line ..."), but it is not, it is discrete and can be faithfully represented by a _discrete_ Fourier transform (which folds at Fs/2). The hiighest frequency in the DFT is Fs, but we know we shouldn't look at that right? There are _no_ 'higher harmonics' present in the data, only in ones 'artistic' interpretation for display purposes. >>> >>> If it looks jagged, that is because in reality sampled data really is! The problem really arises because you do not know how to fill in the space between data samples. You can interpolate (or not). If you interpolate you are making a statement about the model underlying the data and have just carried out a fitting exercise. Fitting is NOT raw data presentation. If you just plot data values you make no assumption about what should join the data, no model has been fit to the data. Every scientist should know the difference between a histogram and a continuous distribution and not be fooled by the vertical lines at the histogram boundaries (which is what you show in a pixel image). >>> >>> The choice is yours, in one case you faithfully show unadulterated sampled data (the histogram looks less 'pretty' than a curve) or you fit a model and interpolate. The trouble with the latter is that the model is probably wrong and you hide the defects in the data (e.g. camera pixel size) from the keen eyed reviewer... Of course if the data points are really close together, the myopic reviewer can't see defects in you data :-) ! From Guy's reasoning, it would be impossible to represent any digitally sampled data because you are always pixelating a continuous function (all pictures get mad up of little squares -the printer dumps blobs of ink etc). So, where does the pixelation become acceptable? This is now aesthetic and has nothing to do with science or mathematics (those with perfect vision will always see discretization 'artifacts' more easily) . >>> >>> Cheers Mark >>> >>> On 16/04/2012, at 3:31 AM, Guy Cox wrote: >>> >>>> ***** >>>> To join, leave or search the confocal microscopy listserv, go to: >>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>> ***** >>>> >>>> OK, having slept on it, I now feel that just maybe I can explain what this is all about. If only the list would let us include pictures it would be much easier! >>>> >>>> Let's assume we have a digital image, from any source, consisting of pixels with a spacing s. The smallest spacing we can resolve in this image is 2s, and this will correspond, in frequency space, with a frequency f. f represents the bandpass limit of this system, no higher frequencies can be passed. Now imagine we have a row of pixels containing the following values: >>>> >>>> 255 0 255 0 255 0 255 0 255 >>>> >>>> If we represent these pixels by little squares, we'll have something like a chessboard. Taking a line along this chessboard will give us a square wave. Now this square wave cannot be represented within the bandpass limit of the system, defined by the frequency f. To represent a square wave we need an infinite series of sine waves f + 3f + 5f +7f ..... To get even a crude approximation to a square wave we need f + 3f - that is a frequency three times higher than the image can contain. >>>> >>>> In other words, we've introduced a whole series of spurious frequencies into our image that not only were not there to start with, they could not possibly have been there. Does this matter? After all, we know they can't be real. It does matter, because we are talking about a visual representation of our data - that's why we drew the little boxes in the first place. Our eyes are very sensitive to edges* and the edges will take over if we let these frequencies come within the bandwidth of our eyes. We will find it very hard to actually see the finest detail in our picture (defined by 2s, remember) because if we enlarge it enough to see this easily we'll also get the edges created by these spurious frequencies. In everyday terms, the pixellation takes over from the picture. >>>> >>>> Note that in all this discussion I have not mentioned microscopes, cameras or anything - we are just talking about a digital image from any source. It applies to confocal, widefield, and electron microscopes, telescopes, X-ray images and your holiday snaps. Coming back to the microscopic world, if we oversample to the point where r, our minimum resolved distance, is substantially greater than 2s, we may not need to enlarge to the point where we see the spurious frequencies. This is probably why some contributors to this discussion have advocated considerable levels of oversampling (though they probably didn't realise this, they just knew they got good pictures that way). But oversampling in fluorescence can be very hard on our specimens. >>>> >>>> "But I'm using a CCD detector so my image is made up of little squares". Yes, you can produce a 'coloured in' picture of your detector that way. I'm assuming the image is actually what you want to see, though, not the detector. >>>> >>>> *Amusingly, the human eye does the same thing to emphasize edges as computer image processing does - it makes the dark side of the edge darker than it is and the light side lighter. >>>> >>>> Guy >>>> >>>> PS. This has doubtless confirmed my reputation among some people as an arrogant bastard. They are probably right, but at least I'm an arrogant bastard who tries to help. It's taken me two hours to write this. |
|
Gordon Scott-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Guys, FWIW, I think much of the dichotomy here is because of what we think the data is. It's neither a true representation of the image, due to quantization and noise, nor it it an image of the sensor, particularly as there are blind 'frames' around every sensor cell, there's no guarantee that any bright or dark spot is central in a sensor cell, or indeed missed completely because of those blind frames. The data is what the sensor 'thinks' it sees. It isn't quite right, but it's the best the sensor can do (per image, anyway). Any reprocessing loses data. If the data you lose are artifacts like noise or harmonics due to the sampling, that's probably fine. If the data you lose are part of the image, that's not so good. Unless there's a very strong reason to do otherwise, keep the original data, unadulterated, warts and all. You can always reprocess it again. Just my two pennyworth. (My opinion, nothing whatever to do with CoolLED) Kind regards, Gordon. -- Gordon Scott --- Electronics and Software Design CoolLED Ltd. www.coolled.com +44-1264-323040 CIL www.cil-uk.co.uk +44-1264-321321 This message has been scanned by MailController - www.MailController.altohiway.com This message and any attachments are strictly confidential and intended solely for the addressee. Any unauthorized use or disclosure, in whole or in part, is prohibited. E-mails are subject to possible alteration. Custom Interconnect Ltd and the sender decline any liability if this message and/or any attachments have been altered, changed or falsified. If you are not the intended recipient of this message, please delete it and notify the sender immediately. CoolLED Ltd is a limited company registered in England. Registered number: 7654232. Registered office: CIL House 48 Charlton road Andover, Hampshire United Kingdom SP103JP. |
| Free forum by Nabble | Edit this page |