To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear all, I will test an Olympus objective on a Nikon Ti ... I know that Olympus does all the chromatic aberration correction at the objective, but am unsure if Nikon distributes this correction between the objective and the tube lens. I would assume that the infinity path is already 'fully' corrected ... but would like to be sure. I appreciate any info on this. Best, Antonio Pereira CID lab i3S/IBMC, Universidade do Porto Room 001.S2B, +351 22 607 49 59 Ext. 6127 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Antonio, I have used Olympus objectives on a Zeiss body before (the 60x silicone oil objective especially) and the mismatch causes chromatic aberration that results in a ~ 0.7 micron offset between ~green and ~red channels. Depending on the application, this is straightforward to correct for using multicolored beads and the SPIM registration plugin for Fiji by Stephan Priebisch and may be worth the advantages of the particular objective you want to use. Whether the degree of aberration will be similar on a Nikon frame, I'm not sure but it should be easy to check. Best, Pavak On Sat, Aug 1, 2020, 7:59 AM Antonio Jose Pereira < [hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear all, > > I will test an Olympus objective on a Nikon Ti ... > I know that Olympus does all the chromatic aberration correction at the > objective, but am unsure if Nikon distributes this correction between the > objective and the tube lens. I would assume that the infinity path is > already 'fully' corrected ... but would like to be sure. I appreciate any > info on this. > > Best, > > Antonio Pereira > CID lab > i3S/IBMC, Universidade do Porto > Room 001.S2B, +351 22 607 49 59 Ext. 6127 > |
|
|
In reply to this post by Antonio Jose Pereira-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Olympus, Nikon and Leica do all corrections in the objective lens. Zeiss does some correction in tube lens. George p.s. in principle (maybe not in practice!) deconvolution software should be able to correct aberrations. p.p.s. deconvolution - Hari Shroff's lab and collaborators have a new fast deconvolution paper, Guo et al 2020 Nature Biotechnology https://www.nature.com/articles/s41587-020-0560-x with matlab/imagej software and code in online supplement (text also mentions it posted to github). Part of the abstract: "Here we describe theoretical and practical advances in algorithm and software design that result in image processing times that are tenfold to several thousand fold faster than with previous methods. First, we show that an ‘unmatched back projector’ accelerates deconvolution relative to the classic Richardson–Lucy algorithm by at least tenfold. Second, three-dimensional image-based registration with a graphics processing unit enhances processing speed 10- to 100-fold over CPU processing." On 8/1/2020 10:58 AM, Antonio Jose Pereira wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear all, > > I will test an Olympus objective on a Nikon Ti ... > I know that Olympus does all the chromatic aberration correction at the objective, but am unsure if Nikon distributes this correction between the objective and the tube lens. I would assume that the infinity path is already 'fully' corrected ... but would like to be sure. I appreciate any info on this. > > Best, > > Antonio Pereira > CID lab > i3S/IBMC, Universidade do Porto > Room 001.S2B, +351 22 607 49 59 Ext. 6127 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I've heard that Leica does a bit of correction on the violet end in the tube lens, but have not verified personally. Zeiss does quite a bit of correction in its tube lens design so you should only use Zeiss with Zeiss. I have interchanged Nikon and Olympus freely without major issue, although since Olympus is designed for a 180mm tube lens and Nikon for 200mm, your field and magnification will be slightly different than usual. Craig On Sat, Aug 1, 2020 at 9:59 AM George McNamara <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Olympus, Nikon and Leica do all corrections in the objective lens. > > Zeiss does some correction in tube lens. > > George > > p.s. in principle (maybe not in practice!) deconvolution software should > be able to correct aberrations. > > p.p.s. deconvolution - Hari Shroff's lab and collaborators have a new > fast deconvolution paper, Guo et al 2020 Nature Biotechnology > https://www.nature.com/articles/s41587-020-0560-x with matlab/imagej > software and code in online supplement (text also mentions it posted to > github). Part of the abstract: > > "Here we describe theoretical and practical advances in algorithm and > software design that result in image processing times that are tenfold > to several thousand fold faster than with previous methods. First, we > show that an ‘unmatched back projector’ accelerates deconvolution > relative to the classic Richardson–Lucy algorithm by at least tenfold. > Second, three-dimensional image-based registration with a graphics > processing unit enhances processing speed 10- to 100-fold over CPU > processing." > > On 8/1/2020 10:58 AM, Antonio Jose Pereira wrote: > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Dear all, > > > > I will test an Olympus objective on a Nikon Ti ... > > I know that Olympus does all the chromatic aberration correction at the > objective, but am unsure if Nikon distributes this correction between the > objective and the tube lens. I would assume that the infinity path is > already 'fully' corrected ... but would like to be sure. I appreciate any > info on this. > > > > Best, > > > > Antonio Pereira > > CID lab > > i3S/IBMC, Universidade do Porto > > Room 001.S2B, +351 22 607 49 59 Ext. 6127 > |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Aside from chromatic distortion, do any of the manufacturers do compensation in the tube lens for other aberrations? Marco On Sat, Aug 1, 2020 at 11:02 Craig Brideau <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > I've heard that Leica does a bit of correction on the violet end in the > tube lens, but have not verified personally. Zeiss does quite a bit of > correction in its tube lens design so you should only use Zeiss with Zeiss. > I have interchanged Nikon and Olympus freely without major issue, although > since Olympus is designed for a 180mm tube lens and Nikon for 200mm, your > field and magnification will be slightly different than usual. > > Craig > > On Sat, Aug 1, 2020 at 9:59 AM George McNamara <[hidden email]> > wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Olympus, Nikon and Leica do all corrections in the objective lens. > > > > Zeiss does some correction in tube lens. > > > > George > > > > p.s. in principle (maybe not in practice!) deconvolution software should > > be able to correct aberrations. > > > > p.p.s. deconvolution - Hari Shroff's lab and collaborators have a new > > fast deconvolution paper, Guo et al 2020 Nature Biotechnology > > https://www.nature.com/articles/s41587-020-0560-x with matlab/imagej > > software and code in online supplement (text also mentions it posted to > > github). Part of the abstract: > > > > "Here we describe theoretical and practical advances in algorithm and > > software design that result in image processing times that are tenfold > > to several thousand fold faster than with previous methods. First, we > > show that an ‘unmatched back projector’ accelerates deconvolution > > relative to the classic Richardson–Lucy algorithm by at least tenfold. > > Second, three-dimensional image-based registration with a graphics > > processing unit enhances processing speed 10- to 100-fold over CPU > > processing." > > > > On 8/1/2020 10:58 AM, Antonio Jose Pereira wrote: > > > ***** > > > To join, leave or search the confocal microscopy listserv, go to: > > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > > Post images on http://www.imgur.com and include the link in your > > posting. > > > ***** > > > > > > Dear all, > > > > > > I will test an Olympus objective on a Nikon Ti ... > > > I know that Olympus does all the chromatic aberration correction at the > > objective, but am unsure if Nikon distributes this correction between the > > objective and the tube lens. I would assume that the infinity path is > > already 'fully' corrected ... but would like to be sure. I appreciate any > > info on this. > > > > > > Best, > > > > > > Antonio Pereira > > > CID lab > > > i3S/IBMC, Universidade do Porto > > > Room 001.S2B, +351 22 607 49 59 Ext. 6127 > > > Marco Sent from my iPhone |
|
Antonio Jose Pereira-2 |
|
|
In reply to this post by Craig Brideau
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Great. Thank you all. The question is answered. Now I wonder why some companies choose to split the correction between the objective and the tube lens. Unless the achieved performance gains from the splitting, we'll then have an infinity path segment which is sub-optimal for injection of e.g. ultra-short pulses. If I'm thinking correctly, the infinity segment is then only 'infinite' at some spectral points after all. Thanks, Antonio ----- Original Message ----- From: "Craig Brideau" <[hidden email]> To: "CONFOCALMICROSCOPY" <[hidden email]> Sent: Saturday, 1 August, 2020 18:56:35 Subject: Re: Aberration correction ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I've heard that Leica does a bit of correction on the violet end in the tube lens, but have not verified personally. Zeiss does quite a bit of correction in its tube lens design so you should only use Zeiss with Zeiss. I have interchanged Nikon and Olympus freely without major issue, although since Olympus is designed for a 180mm tube lens and Nikon for 200mm, your field and magnification will be slightly different than usual. Craig On Sat, Aug 1, 2020 at 9:59 AM George McNamara <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Olympus, Nikon and Leica do all corrections in the objective lens. > > Zeiss does some correction in tube lens. > > George > > p.s. in principle (maybe not in practice!) deconvolution software should > be able to correct aberrations. > > p.p.s. deconvolution - Hari Shroff's lab and collaborators have a new > fast deconvolution paper, Guo et al 2020 Nature Biotechnology > https://www.nature.com/articles/s41587-020-0560-x with matlab/imagej > software and code in online supplement (text also mentions it posted to > github). Part of the abstract: > > "Here we describe theoretical and practical advances in algorithm and > software design that result in image processing times that are tenfold > to several thousand fold faster than with previous methods. First, we > show that an ‘unmatched back projector’ accelerates deconvolution > relative to the classic Richardson–Lucy algorithm by at least tenfold. > Second, three-dimensional image-based registration with a graphics > processing unit enhances processing speed 10- to 100-fold over CPU > processing." > > On 8/1/2020 10:58 AM, Antonio Jose Pereira wrote: > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Dear all, > > > > I will test an Olympus objective on a Nikon Ti ... > > I know that Olympus does all the chromatic aberration correction at the > objective, but am unsure if Nikon distributes this correction between the > objective and the tube lens. I would assume that the infinity path is > already 'fully' corrected ... but would like to be sure. I appreciate any > info on this. > > > > Best, > > > > Antonio Pereira > > CID lab > > i3S/IBMC, Universidade do Porto > > Room 001.S2B, +351 22 607 49 59 Ext. 6127 > |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** You can pre-compensate the laser beam to counter the characteristics of the lens/system. Marco On Sat, Aug 1, 2020 at 12:59 Antonio Jose Pereira < [hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Great. Thank you all. The question is answered. > > Now I wonder why some companies choose to split the correction between the > objective and the tube lens. Unless the achieved performance gains from the > splitting, we'll then have an infinity path segment which is sub-optimal > for injection of e.g. ultra-short pulses. If I'm thinking correctly, the > infinity segment is then only 'infinite' at some spectral points after all. > > Thanks, > Antonio > > ----- Original Message ----- > From: "Craig Brideau" <[hidden email]> > To: "CONFOCALMICROSCOPY" <[hidden email]> > Sent: Saturday, 1 August, 2020 18:56:35 > Subject: Re: Aberration correction > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > I've heard that Leica does a bit of correction on the violet end in the > tube lens, but have not verified personally. Zeiss does quite a bit of > correction in its tube lens design so you should only use Zeiss with Zeiss. > I have interchanged Nikon and Olympus freely without major issue, although > since Olympus is designed for a 180mm tube lens and Nikon for 200mm, your > field and magnification will be slightly different than usual. > > Craig > > On Sat, Aug 1, 2020 at 9:59 AM George McNamara <[hidden email]> > wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Olympus, Nikon and Leica do all corrections in the objective lens. > > > > Zeiss does some correction in tube lens. > > > > George > > > > p.s. in principle (maybe not in practice!) deconvolution software should > > be able to correct aberrations. > > > > p.p.s. deconvolution - Hari Shroff's lab and collaborators have a new > > fast deconvolution paper, Guo et al 2020 Nature Biotechnology > > https://www.nature.com/articles/s41587-020-0560-x with matlab/imagej > > software and code in online supplement (text also mentions it posted to > > github). Part of the abstract: > > > > "Here we describe theoretical and practical advances in algorithm and > > software design that result in image processing times that are tenfold > > to several thousand fold faster than with previous methods. First, we > > show that an ‘unmatched back projector’ accelerates deconvolution > > relative to the classic Richardson–Lucy algorithm by at least tenfold. > > Second, three-dimensional image-based registration with a graphics > > processing unit enhances processing speed 10- to 100-fold over CPU > > processing." > > > > On 8/1/2020 10:58 AM, Antonio Jose Pereira wrote: > > > ***** > > > To join, leave or search the confocal microscopy listserv, go to: > > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > > Post images on http://www.imgur.com and include the link in your > > posting. > > > ***** > > > > > > Dear all, > > > > > > I will test an Olympus objective on a Nikon Ti ... > > > I know that Olympus does all the chromatic aberration correction at the > > objective, but am unsure if Nikon distributes this correction between the > > objective and the tube lens. I would assume that the infinity path is > > already 'fully' corrected ... but would like to be sure. I appreciate any > > info on this. > > > > > > Best, > > > > > > Antonio Pereira > > > CID lab > > > i3S/IBMC, Universidade do Porto > > > Room 001.S2B, +351 22 607 49 59 Ext. 6127 > > > Marco Sent from my iPhone |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Without having an actual model of the lenses it is hard to say for certain, but I can say from personal experience in engineering optical systems, it is very hard and expensive to have a fully corrected single optical element, especially one that works across the broad array of wavelengths and resolutions encountered on a microscope. The moment you fix one aberration, you exacerbate another as everything is kind of interdependent. Therefore, quite often it is much easier to get things close to perfect within each lens system of the microscope, and then simply ensure that the slight remaining aberrations within each lens system have opposite signs completing the correction across a wider array of wavelengths. For example, even with something as simple as an achromatic Gallilean beam expander, you will run into this lens-pair dependent effect. Even though as a stand-alone lens, you will find that most vendor's lenses tend to perform close to one another, they will all have different residual aberrations (due to different design wavelengths, glass formulation/thickness, etc.). Therefore, by carefully combining the right lens pair such that the signs of their residual aberrations are opposite one another, you can get your optical system to well out-perform the specs on either individual component. As such, since the microscope manufacturers know the composition and residual aberrations of their own optical systems, they have the advantage of making the objective and tube lens complimentary in this way to further increase the performance above that of just the objective or tube lens alone (or at least ensure the residual aberrations are not additive). Not saying this is what they do, but it would definitely make sense if they did. This would also mean that mixing and matching could lead to additive residual aberrations making the system underperform, although this underperformance should be slight, since the objective and tube lens are already extensively corrected on their own. -Ben Smith On Sat, Aug 1, 2020 at 1:04 PM Marc Reinig <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > You can pre-compensate the laser beam to counter the characteristics of the > lens/system. > > Marco > > On Sat, Aug 1, 2020 at 12:59 Antonio Jose Pereira < > [hidden email]> wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Great. Thank you all. The question is answered. > > > > Now I wonder why some companies choose to split the correction between > the > > objective and the tube lens. Unless the achieved performance gains from > the > > splitting, we'll then have an infinity path segment which is sub-optimal > > for injection of e.g. ultra-short pulses. If I'm thinking correctly, the > > infinity segment is then only 'infinite' at some spectral points after > all. > > > > Thanks, > > Antonio > > > > ----- Original Message ----- > > From: "Craig Brideau" <[hidden email]> > > To: "CONFOCALMICROSCOPY" <[hidden email]> > > Sent: Saturday, 1 August, 2020 18:56:35 > > Subject: Re: Aberration correction > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > I've heard that Leica does a bit of correction on the violet end in the > > tube lens, but have not verified personally. Zeiss does quite a bit of > > correction in its tube lens design so you should only use Zeiss with > Zeiss. > > I have interchanged Nikon and Olympus freely without major issue, > although > > since Olympus is designed for a 180mm tube lens and Nikon for 200mm, your > > field and magnification will be slightly different than usual. > > > > Craig > > > > On Sat, Aug 1, 2020 at 9:59 AM George McNamara < > [hidden email]> > > wrote: > > > > > ***** > > > To join, leave or search the confocal microscopy listserv, go to: > > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > > Post images on http://www.imgur.com and include the link in your > > posting. > > > ***** > > > > > > Olympus, Nikon and Leica do all corrections in the objective lens. > > > > > > Zeiss does some correction in tube lens. > > > > > > George > > > > > > p.s. in principle (maybe not in practice!) deconvolution software > should > > > be able to correct aberrations. > > > > > > p.p.s. deconvolution - Hari Shroff's lab and collaborators have a new > > > fast deconvolution paper, Guo et al 2020 Nature Biotechnology > > > https://www.nature.com/articles/s41587-020-0560-x with matlab/imagej > > > software and code in online supplement (text also mentions it posted to > > > github). Part of the abstract: > > > > > > "Here we describe theoretical and practical advances in algorithm and > > > software design that result in image processing times that are tenfold > > > to several thousand fold faster than with previous methods. First, we > > > show that an ‘unmatched back projector’ accelerates deconvolution > > > relative to the classic Richardson–Lucy algorithm by at least tenfold. > > > Second, three-dimensional image-based registration with a graphics > > > processing unit enhances processing speed 10- to 100-fold over CPU > > > processing." > > > > > > On 8/1/2020 10:58 AM, Antonio Jose Pereira wrote: > > > > ***** > > > > To join, leave or search the confocal microscopy listserv, go to: > > > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > > > Post images on http://www.imgur.com and include the link in your > > > posting. > > > > ***** > > > > > > > > Dear all, > > > > > > > > I will test an Olympus objective on a Nikon Ti ... > > > > I know that Olympus does all the chromatic aberration correction at > the > > > objective, but am unsure if Nikon distributes this correction between > the > > > objective and the tube lens. I would assume that the infinity path is > > > already 'fully' corrected ... but would like to be sure. I appreciate > any > > > info on this. > > > > > > > > Best, > > > > > > > > Antonio Pereira > > > > CID lab > > > > i3S/IBMC, Universidade do Porto > > > > Room 001.S2B, +351 22 607 49 59 Ext. 6127 > > > > > > -- > Marco > > Sent from my iPhone > -- Benjamin E. Smith, Ph. D. Imaging Specialist, Vision Science University of California, Berkeley 195 Life Sciences Addition Berkeley, CA 94720-3200 Tel (510) 642-9712 Fax (510) 643-6791 e-mail: [hidden email] https://vision.berkeley.edu/faculty/core-grants-nei/core-grant-microscopic-imaging/ |
|
Ciaran Butler-Hallissey |
|
|
In reply to this post by Antonio Jose Pereira-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Has anyone ever been routinely provided (or asked vendors) for ranges of "reasonable" chromatic abberation per objective? For instance, after installation or service when a system is aligned correctly. Best, Ciaran Cell Imaging and Cytometry, Turku Bioscience, Website: https://bioscience.fi/cic |