Alexa Fluor 488 / solubility problem

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
I'm seeking advice on how to dissolve the Alexa Fluor 488 phenylmercury dye from Molecular Probes?

I am finding it impossible. I have tried three different solvents: DMSO, DMF, and 100 mM HEPES pH 7.5. I have tried adding from 100-1000 ul of solvent to the vials (which contain 1 mg dye according to the manufacturer). I have vortexed, warmed, and waited while trying with a total of six different vials from two lots. In all cases the majority of the dye remains as solids at the bottom of the vial. In addition, the supernatant does not appear to be dissolved dye. I am unable to remove it with either a spin column, a 3000 MWCO, or even a 10000 MWCO dialysis lasting many days. I have heard that the dye molecules associate by stacking, making for a seemingly high molecular weight. Molecular Probes Tech Support claims ignorance of any problems.

Ideally I would like to be able to add 100 ul of solvent to 1 mg of dye (to obtain an ~11 mM solution, which I can then add in 1/10 volume to my sample, to have dye in ~10-fold molar excess over my ~100 uM protein). I would like to keep the solvent (if its DMF or DMSO) below 10% final concentration to minimize denaturation. I have seen reports in the literature where other Alexa 488 series dyes have supposedly been prepared as 10 mg/ml stock solutions in DMF.

Norman


__________________________________________________


Norman Watts, Ph. D.

National Institutes of Health

50 South Drive, Rm. 1509

Bethesda, MD 20892-8025

Phone: (301) 402-3418

Fax: (301) 480-7629


Peng Xi-2 Peng Xi-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Detector question -- APD

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi,

    Anybody know which company produces high-quality  Avalanche photodiode (APD)? I am planning to use it to detect 800nm signal. I prefer high sensitivity, low noise, analog one used in confocal reflective devices. Thank you for your advice!

 

Best regards,
Peng Xi
Associate Professor
Institute for Laser Medicine and Biophotonics
Shanghai Jiao Tong University
800 Dongchuan Rd.
Shanghai 200240, China
Tel: (86) 21-3420-4076
http://biophotonics.sjtu.edu.cn/

 

Alberto Diaspro Alberto Diaspro
Reply | Threaded
Open this post in threaded view
♦
♦
|

mCherry

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Ciao
Is there anyone working with 2PE and mCherry? In case, do you have  
excitation spectrum, a real one?
All  the best
Alby

----------------------------------------------------
"Water slowly flowed under the sky" (Cesare Pavese)
-----------------------------------------------------
  Alberto Diaspro, EBSA President-Elect, MicroScoBIO LAMBS-IFOM,  
Department of Physics, University of Genoa, Via Dodecaneso 33, 16146  
Genoa, Italy - fax +39-010314218 - tel +39 0103536426/309; URLs: www.lambs.it
;
EBSA is Biophysics in Europe - European Biophysical Societies'  
Association www.ebsa.org
  ----------------------------------------------
Dale Callaham Dale Callaham
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Detector question -- APD

In reply to this post by Peng Xi-2
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Have a look at hamamatsu? I'm sure they are high quality.
http://sales.hamamatsu.com/en/products/solid-state-division/si-photodiode-series/si-apd.php

I have no connection with Hamamatsu.

Dale

Peng Xi wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi,
>
>     Anybody know which company produces high-quality  Avalanche
> photodiode (APD)? I am planning to use it to detect 800nm signal. I
> prefer high sensitivity, low noise, analog one used in confocal
> reflective devices. Thank you for your advice!
>
>  
>
> Best regards,
> Peng Xi
> Associate Professor
> Institute for Laser Medicine and Biophotonics
> Shanghai Jiao Tong University
> 800 Dongchuan Rd.
> Shanghai 200240, China
> Tel: (86) 21-3420-4076
> http://biophotonics.sjtu.edu.cn/
>
>  
>
Dale Callaham Dale Callaham
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Detector question -- APD - a bit more info

In reply to this post by Peng Xi-2
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I also noticed that Hamamatsu has some applications literature that
might be of interest:

for APDs:
http://sales.hamamatsu.com/assets/applications/SSD/Characteristics_and_use_of_SI_APD.pdf

and also other detectors:
http://sales.hamamatsu.com/en/support/technical-notes.php

Again, I have no connection with Hamamatsu, but thank them for making
good technical information available!

Dale

Peng Xi wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi,
>
>     Anybody know which company produces high-quality  Avalanche
> photodiode (APD)? I am planning to use it to detect 800nm signal. I
> prefer high sensitivity, low noise, analog one used in confocal
> reflective devices. Thank you for your advice!
>
>  
>
> Best regards,
> Peng Xi
> Associate Professor
> Institute for Laser Medicine and Biophotonics
> Shanghai Jiao Tong University
> 800 Dongchuan Rd.
> Shanghai 200240, China
> Tel: (86) 21-3420-4076
> http://biophotonics.sjtu.edu.cn/
>
>  
>
Michael Weber-4 Michael Weber-4
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: mCherry

In reply to this post by Alberto Diaspro
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear Alby,

check the Svoboda paper: PLOS Biology in 2006 vol. 4 (11) pp. e370 (Rapid
redistribution of synaptic PSD-95 in the neocortex in vivo.). They used a
fixed wavelength 1030 nm laser for mCherry (110 mW in bfp). We tried ~1050
nm on Spectraphysics MaiTai laser, but power was too low (while still
getting nice images with 559 nm single photon). An Ultra II should give a
bit more power in this range, however, still not much (200 mW at laser
head). So either the mentioned laser from Amplitude Systems, or maybe this
OPO unit.

Michael


> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Ciao
> Is there anyone working with 2PE and mCherry? In case, do you have
> excitation spectrum, a real one?
> All  the best
> Alby
>
> ----------------------------------------------------
> "Water slowly flowed under the sky" (Cesare Pavese)
> -----------------------------------------------------
>   Alberto Diaspro, EBSA President-Elect, MicroScoBIO LAMBS-IFOM,
> Department of Physics, University of Genoa, Via Dodecaneso 33, 16146
> Genoa, Italy - fax +39-010314218 - tel +39 0103536426/309; URLs:
> www.lambs.it
> ;
> EBSA is Biophysics in Europe - European Biophysical Societies'
> Association www.ebsa.org
>   ----------------------------------------------
Peng Xi-2 Peng Xi-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Detector question -- APD

In reply to this post by Dale Callaham
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Dale,
    Yes, I contacted them in the first place, as they have the best PMT
(in my opinion). But the APD they recommend me is really not good
(C5460) -- 2-3 times weaker signal than the PMT. I think APD should
perform better than PMT in 800nm region, and that's why many reflective
confocal setup uses APD (well, also APD should be cheaper).

Best regards,
Peng Xi
Associate Professor
Institute for Laser Medicine and Biophotonics
Shanghai Jiao Tong University
800 Dongchuan Rd.
Shanghai 200240, China
Tel: (86) 21-3420-4076
http://biophotonics.sjtu.edu.cn/



Dale Callaham wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Have a look at hamamatsu? I'm sure they are high quality.
> http://sales.hamamatsu.com/en/products/solid-state-division/si-photodiode-series/si-apd.php 
>
>
> I have no connection with Hamamatsu.
>
> Dale
>
> Peng Xi wrote:
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>> Hi,
>>
>>     Anybody know which company produces high-quality  Avalanche
>> photodiode (APD)? I am planning to use it to detect 800nm signal. I
>> prefer high sensitivity, low noise, analog one used in confocal
>> reflective devices. Thank you for your advice!
>>
>>  
>>
>> Best regards,
>> Peng Xi
>> Associate Professor
>> Institute for Laser Medicine and Biophotonics
>> Shanghai Jiao Tong University
>> 800 Dongchuan Rd.
>> Shanghai 200240, China
>> Tel: (86) 21-3420-4076
>> http://biophotonics.sjtu.edu.cn/
>>
>>  
>>
Dale Callaham Dale Callaham
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Detector question -- APD

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Professor Xi,

I think that the main reason APDs are used in that application is that
they are fast detectors, not affected by the potential high ambient
light exposure and there are plenty of photons available so sensitivity
is less an issue.  Below I am pasting in some text from a Hamamatsu paper:
http://sales.hamamatsu.com/assets/applications/Combined/lightlevelnoise.pdf

I don't know if this is any help. I think that ~0.5A/W is all that
silicon is capable of yielding, and then it is a matter of signal,
speed, or noise level, or damage by high ambient light.

But I have to admit I do not work with this in detail
and there may be other issues.


Dale


The avalanche photodiode (Figure 3) has features of both the photodiode
and the photomultiplier tube. It is a solid-state device that generates
electron-hole pairs upon exposure to light. A reverse bias of 100 to
1000 V is placed on the PN junction, creating an internal electric field
large enough to accelerate the electrons. When the electrons collide
with the crystal lattice, they generate electron-hole pairs. This
process cascades, amplifying the initial signal by a factor of 50 or more.

The statistical fluctuations in the number of collisions
and the yield of electron-hole pairs create more noise than
the dynodes in a photomultiplier tube but less noise than
a photodiode’s external amplifier. Thus, avalanche photodiodes
are often used when light levels are too high for photomultiplier
tubes but not high enough for photodiodes.

Avalanche photodiodes are inadequate for some applications,
such as those in which the lifetime of the photon source is short, a
measurement must be made rapidly, few photons arrive at the detector or
the photons are spread over a large area. These situations will usually
require a photomultiplier tube.

Peng Xi wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi Dale, Yes, I contacted them in the first place, as they have the
> best PMT (in my opinion). But the APD they recommend me is really not
> good (C5460) -- 2-3 times weaker signal than the PMT. I think APD
> should perform better than PMT in 800nm region, and that's why many
> reflective confocal setup uses APD (well, also APD should be
> cheaper).
>
> Best regards, Peng Xi Associate Professor Institute for Laser
> Medicine and Biophotonics Shanghai Jiao Tong University 800 Dongchuan
> Rd. Shanghai 200240, China Tel: (86) 21-3420-4076
> http://biophotonics.sjtu.edu.cn/
>
>
>
> Dale Callaham wrote:
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>> Have a look at hamamatsu? I'm sure they are high quality.
>> http://sales.hamamatsu.com/en/products/solid-state-division/si-photodiode-series/si-apd.php
>>
>>
>>
>> I have no connection with Hamamatsu.
>>
>> Dale
>>
>> Peng Xi wrote:
>>> Search the CONFOCAL archive at
>>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>>
>>> Hi,
>>>
>>> Anybody know which company produces high-quality  Avalanche
>>> photodiode (APD)? I am planning to use it to detect 800nm signal.
>>> I prefer high sensitivity, low noise, analog one used in confocal
>>>  reflective devices. Thank you for your advice!
>>>
>>>
>>>
>>> Best regards, Peng Xi Associate Professor Institute for Laser
>>> Medicine and Biophotonics Shanghai Jiao Tong University 800
>>> Dongchuan Rd. Shanghai 200240, China Tel: (86) 21-3420-4076
>>> http://biophotonics.sjtu.edu.cn/
>>>
>>>
>>>
James Pawley James Pawley
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Detector question -- APD

In reply to this post by Peng Xi-2
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi all,

I confess that I know nothing of the specific APDs in question.

However some years ago, I was very interested in APDs in general and,
at that time, those that worked in the analog mode, had to operate at
low gain (10- 100x) so that they did not run away. In this mode they
had very high multipicative noise* and in addition, at this low field
strength, the most likely amplification for an individual
photoelectron (PE) was zero. (i.e., Although the photon was absorbed
and the PE produced, the local field was insufficiently large to make
the PE initiate a cascade). All the PE with a gain of zero
essentially reduced the effective QE to a level much below that of
the silicon itself.

If you increased the acceleration field enough that most PE did
initiate a cascade, this produced a gain of maybe 10 million (+/- a
lot, depending on where in the depletion zone the PE was produced.).
However, you then had to quench this process (either passively, by
putting a resistor in series with the diode, so that the current flow
of the discharge reduced the applied voltage and terminated the
discharge, or by using active circuits to do the same thing but with
a shorter time-constant.) to prevent the diode from overheating.

This meant that you could only operate in a pulse-counting mode. If
the maximum throughput of the pulse-counting circuit was 30 MHz, then
even counting 3 photons in a 1 microsecond pixel time would lead to a
10% chance of having missed one more pulse because of pulse pileup: a
30% error that would become WORSE as the count-rate increased. Not
very useful for normal confocal.

At that time, you could only get a high effective QE if you used a
high field strength, and is you did this you had to use pulse
counting.

It would be nice to know away around this problem but I have yet to
hear of one.

Jim P.

*The extreme end of this trend is a single stage of the gain register
in an EM-CCD with a gain of 1.01, a situation that produces a
multiplicative noise almost exactly as large as the Poisson Noise
from the same signal and effectively halves the QE.


               **********************************************
Prof. James B. Pawley,                          Ph.  608-263-3147
Room 223, Zoology Research Building,              
FAX  608-265-5315
1117 Johnson Ave., Madison, WI, 53706  
[hidden email]
3D Microscopy of Living Cells Course, June 14-26, 2008, UBC, Vancouver Canada
Info: http://www.3dcourse.ubc.ca/                 Applications still
being accepted
               "If it ain't diffraction, it must be statistics." Anon.


>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>Hi Dale,
>    Yes, I contacted them in the first place, as they have the best
>PMT (in my opinion). But the APD they recommend me is really not
>good (C5460) -- 2-3 times weaker signal than the PMT. I think APD
>should perform better than PMT in 800nm region, and that's why many
>reflective confocal setup uses APD (well, also APD should be
>cheaper).
>
>Best regards,
>Peng Xi Associate Professor
>Institute for Laser Medicine and Biophotonics
>Shanghai Jiao Tong University
>800 Dongchuan Rd.
>Shanghai 200240, China
>Tel: (86) 21-3420-4076
>http://biophotonics.sjtu.edu.cn/
>
>
>
>Dale Callaham wrote:
>>Search the CONFOCAL archive at
>>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>>Have a look at hamamatsu? I'm sure they are high quality.
>>http://sales.hamamatsu.com/en/products/solid-state-division/si-photodiode-series/si-apd.php
>>
>>I have no connection with Hamamatsu.
>>
>>Dale
>>
>>Peng Xi wrote:
>>>Search the CONFOCAL archive at
>>>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>>
>>>Hi,
>>>
>>>     Anybody know which company produces high-quality  Avalanche
>>>photodiode (APD)? I am planning to use it to detect 800nm signal.
>>>I prefer high sensitivity, low noise, analog one used in confocal
>>>reflective devices. Thank you for your advice!
>>>
>>>
>>>Best regards,
>>>Peng Xi
>>>Associate Professor
>>>Institute for Laser Medicine and Biophotonics
>>>Shanghai Jiao Tong University
>>>800 Dongchuan Rd.
>>>Shanghai 200240, China
>>>Tel: (86) 21-3420-4076
>>>http://biophotonics.sjtu.edu.cn/
>>>
>>>


--
Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Detector question -- APD

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Re: Detector question -- APD
I am surprised that Leica haven't replied to this thread since (to my
knowledge) they are the only major manufacturer using APD detectors.
I have used their APD-equipped STED system and can comment that
the APD is much more sensitive than PMT detection (in the wavelength
range 650-730nm) but is very sensitive to excess light - with a bright
sample the software can shut down the detector to prevent damage
halfway through collecting an image, which can be annoying.  But in
the end you live with this since you need the sensitivity.  As I understand
it they are working in photon-counting mode, but I must assume that
it's at a much higher rate than Jim suggests. 
 
                                                                                             Guy
 
 
Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
     http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
http://www.guycox.net


From: Confocal Microscopy List on behalf of James Pawley
Sent: Fri 08/04/04 8:42 AM
To: [hidden email]
Subject: Re: Detector question -- APD

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi all,

I confess that I know nothing of the specific APDs in question.

However some years ago, I was very interested in APDs in general and,
at that time, those that worked in the analog mode, had to operate at
low gain (10- 100x) so that they did not run away. In this mode they
had very high multipicative noise* and in addition, at this low field
strength, the most likely amplification for an individual
photoelectron (PE) was zero. (i.e., Although the photon was absorbed
and the PE produced, the local field was insufficiently large to make
the PE initiate a cascade). All the PE with a gain of zero
essentially reduced the effective QE to a level much below that of
the silicon itself.

If you increased the acceleration field enough that most PE did
initiate a cascade, this produced a gain of maybe 10 million (+/- a
lot, depending on where in the depletion zone the PE was produced.).
However, you then had to quench this process (either passively, by
putting a resistor in series with the diode, so that the current flow
of the discharge reduced the applied voltage and terminated the
discharge, or by using active circuits to do the same thing but with
a shorter time-constant.) to prevent the diode from overheating.

This meant that you could only operate in a pulse-counting mode. If
the maximum throughput of the pulse-counting circuit was 30 MHz, then
even counting 3 photons in a 1 microsecond pixel time would lead to a
10% chance of having missed one more pulse because of pulse pileup: a
30% error that would become WORSE as the count-rate increased. Not
very useful for normal confocal.

At that time, you could only get a high effective QE if you used a
high field strength, and is you did this you had to use pulse
counting.

It would be nice to know away around this problem but I have yet to
hear of one.

Jim P.

*The extreme end of this trend is a single stage of the gain register
in an EM-CCD with a gain of 1.01, a situation that produces a
multiplicative noise almost exactly as large as the Poisson Noise
from the same signal and effectively halves the QE.


               **********************************************
Prof. James B. Pawley,                                      Ph.  608-263-3147
Room 223, Zoology Research Building,              
FAX  608-265-5315
1117 Johnson Ave., Madison, WI, 53706 
[hidden email]
3D Microscopy of Living Cells Course, June 14-26, 2008, UBC, Vancouver Canada
Info: http://www.3dcourse.ubc.ca/                Applications still
being accepted
               "If it ain't diffraction, it must be statistics." Anon.


>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>Hi Dale,
>    Yes, I contacted them in the first place, as they have the best
>PMT (in my opinion). But the APD they recommend me is really not
>good (C5460) -- 2-3 times weaker signal than the PMT. I think APD
>should perform better than PMT in 800nm region, and that's why many
>reflective confocal setup uses APD (well, also APD should be
>cheaper).
>
>Best regards,
>Peng Xi Associate Professor
>Institute for Laser Medicine and Biophotonics
>Shanghai Jiao Tong University
>800 Dongchuan Rd.
>Shanghai 200240, China
>Tel: (86) 21-3420-4076
>http://biophotonics.sjtu.edu.cn/
>
>
>
>Dale Callaham wrote:
>>Search the CONFOCAL archive at
>>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>>Have a look at hamamatsu? I'm sure they are high quality.
>>http://sales.hamamatsu.com/en/products/solid-state-division/si-photodiode-series/si-apd.php
>>
>>I have no connection with Hamamatsu.
>>
>>Dale
>>
>>Peng Xi wrote:
>>>Search the CONFOCAL archive at
>>>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>>
>>>Hi,
>>>
>>>     Anybody know which company produces high-quality  Avalanche
>>>photodiode (APD)? I am planning to use it to detect 800nm signal.
>>>I prefer high sensitivity, low noise, analog one used in confocal
>>>reflective devices. Thank you for your advice!
>>>
>>>
>>>Best regards,
>>>Peng Xi
>>>Associate Professor
>>>Institute for Laser Medicine and Biophotonics
>>>Shanghai Jiao Tong University
>>>800 Dongchuan Rd.
>>>Shanghai 200240, China
>>>Tel: (86) 21-3420-4076
>>>http://biophotonics.sjtu.edu.cn/
>>>
>>>


--

Csucs Gabor Csucs Gabor
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Detector question -- APD

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear Guy,

Just to make it more precise: You can do APD-imaging also with a
Zeiss-confocal if you have the ConfoCor3 FCS/FCCS setup. IT works very
nicely and gives you indeed a huge signal increase.

Cheers    Gabor

--
Gabor Csucs
Light Microscopy Centre, ETH Zurich
Schafmattstrasse 18, HPM F16
CH-8093, Zurich, Switzerland

Web: www.lmc.ethz.ch
Phone: +41 44 633 6221
Fax: +41 44 632 1298
e-mail: [hidden email]
James Pawley James Pawley
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Detector question -- APD

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>Dear Guy,
>
>Just to make it more precise: You can do APD-imaging also with a
>Zeiss-confocal if you have the ConfoCor3 FCS/FCCS setup. IT works
>very nicely and gives you indeed a huge signal increase.
>
>Cheers    Gabor
>
>--
>Gabor Csucs Light Microscopy Centre, ETH Zurich
>Schafmattstrasse 18, HPM F16 CH-8093, Zurich, Switzerland
>
>Web: www.lmc.ethz.ch
>Phone: +41 44 633 6221
>Fax: +41 44 632 1298
>e-mail: [hidden email]

I agree that ConFoCOr seems an appropriate use: the signal level is
very low indeed. And the contrast is basically the presence or
absence of a single fluorophor in the spot.

The STED also is noted for low signal levels as most of the
excitations are "entrained" by stimulated emission to go away from
the detector.  On the other hand, even with a 30MHz counter you can
conceivably count up to about 20 pulses in a microsecond, before the
pileup losses become too obvious to the eye, while a faster circuit
or a longer pixel-dwell time allow you to count proportionally more.

Cheers,

Jim P.
--
               **********************************************
Prof. James B. Pawley,                          Ph.  608-263-3147
Room 223, Zoology Research Building,              
FAX  608-265-5315
1117 Johnson Ave., Madison, WI, 53706  
[hidden email]
3D Microscopy of Living Cells Course, June 14-26, 2008, UBC, Vancouver Canada
Info: http://www.3dcourse.ubc.ca/             Applications still being accepted
               "If it ain't diffraction, it must be statistics." Anon.
Quoc Thang Nguyen Quoc Thang Nguyen
Reply | Threaded
Open this post in threaded view
♦
♦
|

Pixel integrator

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear all,

Does anyone know of a stand-alone multiphase pixel integrator system
suitable for  custom-built laser scanning scopes? Sometime ago, Cairn
Research in the UK was marketing such a system but it seems to have
been discontinued.

Thanks.



Quoc Thang NGUYEN, Ph.D.
Assistant Project Scientist
Physics Department, UCSD
e-mail: [hidden email]
Cell Phone: (949) 246-8143
Fax: (858) 534-7697