						Farid Jalali | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello all, Is anyone aware of a suitable alternative immersion media for a water immersion lens? This lab is doing experiments at room temperature and the scope is not enclosed therefore no humidity control. Thanks very much all. Farid Jalali  | 
			
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Yes, you can get a silicone (I think) immersion fluid with the same refractive index as water. I don't know a supplier off hand but I'm sure someone here does. Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________ Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Farid Jalali Sent: Friday, 22 July 2011 2:43 PM To: [hidden email] Subject: Alternative immersion media for a water immersion lens ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello all, Is anyone aware of a suitable alternative immersion media for a water immersion lens? This lab is doing experiments at room temperature and the scope is not enclosed therefore no humidity control. Thanks very much all. Farid Jalali ----- No virus found in this message. Checked by AVG - www.avg.com Version: 10.0.1390 / Virus Database: 1518/3779 - Release Date: 07/21/11  | 
			
 
	
	 | 
						George McNamara | 
| 
					
		 | 
				
					
	 
				In reply to this post by Farid Jalali
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Zeiss Immersol W immersion oil with refractive index 1.33. Should be available through the Zeiss micro-shop web site. On 7/22/2011 12:42 AM, Farid Jalali wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello all, > Is anyone aware of a suitable alternative immersion media for a water > immersion lens? This lab is doing experiments at room temperature and the > scope is not enclosed therefore no humidity control. > Thanks very much all. > > Farid Jalali > > -- George McNamara, PhD Analytical Imaging Core Facility University of Miami  | 
			
 
	
	 | 
						Cameron Nowell | 
| 
					
		 | 
				
					
	 
				In reply to this post by Farid Jalali
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Farid, For small volumes we use the Zeiss Immersol W. But for large scale stuff like using the large 20x water objective on our multiphoton system we use a 1:5 dilution of ultrasound gel (make sure you use the clear one, not the blue tinted stuff). Cheers Cam -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Farid Jalali Sent: Friday, 22 July 2011 2:43 PM To: [hidden email] Subject: Alternative immersion media for a water immersion lens ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello all, Is anyone aware of a suitable alternative immersion media for a water immersion lens? This lab is doing experiments at room temperature and the scope is not enclosed therefore no humidity control. Thanks very much all. Farid Jalali This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research Ltd does not waive any rights if you have received this communication in error. The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research Ltd.  | 
			
 
	
	 | 
						Oshel, Philip Eugene | 
| 
					
		 | 
				
					
	 
				In reply to this post by Farid Jalali
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Cargille makes water-RI immersion oils for both 37 deg C and 25 deg C. (The 37 deg is 1.3350 and the 25 deg is 1.330 - by their labels.) These are in their "AAA" series of immersion oils. Phil >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Hello all, >Is anyone aware of a suitable alternative immersion media for a water >immersion lens? This lab is doing experiments at room temperature and the >scope is not enclosed therefore no humidity control. >Thanks very much all. > >Farid Jalali -- Philip Oshel Microscopy Facility Supervisor Biology Department 024C Brooks Hall Central Michigan University Mt. Pleasant, MI 48859 (989) 774-3576  | 
			
 
	
	 | 
						Steffen Dietzel | 
| 
					
		 | 
				
					
	 
				In reply to this post by Cameron Nowell
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi there, I have used 1:5 ultrasound gel too in the past. The Ri is not identical to water but close enough for many applications. Just make sure you use water for dilution. With NaCl solution, somehow the gel structure breaks down and you get an aqueous liquid with a precipitate, at least with our ultrasound gel. I don't know though what the 1:5 dilution in water does to exposed tissues in terms of physiology, as opposed to superfusion with buffer. Did anybody investigate? Steffen On 22.07.2011 09:08, Cameron Nowell wrote: > > Hi Farid, > > For small volumes we use the Zeiss Immersol W. But for large scale stuff > like using the large 20x water objective on our multiphoton system we > use a 1:5 dilution of ultrasound gel (make sure you use the clear one, > not the blue tinted stuff). > -- ------------------------------------------------------------ Steffen Dietzel, PD Dr. rer. nat Ludwig-Maximilians-Universität München Walter-Brendel-Zentrum für experimentelle Medizin (WBex) Head of light microscopy Mail room: Marchioninistr. 15, D-81377 München Building location: Marchioninistr. 27, München-Großhadern  | 
			
 
	
	 | 
						Farid Jalali | 
| 
					
		 | 
				
					
	 
				In reply to this post by Farid Jalali
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello All. Thank you very much to everyone for your replies and insightful suggestions. It never ceases to amaze me what a wonderful community this is. Have a great day all. Cheers Farid On Thu, Jul 21, 2011 at 9:42 PM, Farid Jalali <[hidden email]> wrote: > Hello all, > Is anyone aware of a suitable alternative immersion media for a water > immersion lens? This lab is doing experiments at room temperature and the > scope is not enclosed therefore no humidity control. > Thanks very much all. > > Farid Jalali >  | 
			
 
	
	 | 
						Stephen Firth (Med) | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** The other way to do it is a continual water source. We have the Leica 63x with computer controlled water pump (no commercial interest), but also use a lab-made system using a pipette, reservoir and pump (designed by Steve Cody). We also use the Zeiss immersol and Cargille water RI oils but find they dry over time and the quality of the image deteriorates. The water pump also allow imaging multipoint experiments where there is considerable movement of the objective. Cheers, Stephen On 23 July 2011 03:19, Farid Jalali <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello All. Thank you very much to everyone for your replies and insightful > suggestions. It never ceases to amaze me what a wonderful community this > is. > Have a great day all. > Cheers > Farid > > On Thu, Jul 21, 2011 at 9:42 PM, Farid Jalali <[hidden email]> > wrote: > > > Hello all, > > Is anyone aware of a suitable alternative immersion media for a water > > immersion lens? This lab is doing experiments at room temperature and the > > scope is not enclosed therefore no humidity control. > > Thanks very much all. > > > > Farid Jalali > > > -- Stephen Firth Manager of Advanced Optical Microscopy Monash Micro Imaging Monash University Room G58 Building 75 Phone 990 55612 Mob 0400 695 425  | 
			
 
	
	 | 
						Cameron Nowell | 
| 
					
		 | 
				
					
	 
				In reply to this post by Steffen Dietzel
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Steffen For any live animal work we dilute in PBS or 0.9% saline without any problems. Cheers Cam Sent from my iPhone On 23/07/2011, at 12:58 AM, "Steffen Dietzel" <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi there, > > I have used 1:5 ultrasound gel too in the past. The Ri is not identical to water but close enough for many applications. Just make sure you use water for dilution. With NaCl solution, somehow the gel structure breaks down and you get an aqueous liquid with a precipitate, at least with our ultrasound gel. > > I don't know though what the 1:5 dilution in water does to exposed tissues in terms of physiology, as opposed to superfusion with buffer. Did anybody investigate? > > Steffen > > > On 22.07.2011 09:08, Cameron Nowell wrote: >> >> Hi Farid, >> >> For small volumes we use the Zeiss Immersol W. But for large scale stuff >> like using the large 20x water objective on our multiphoton system we >> use a 1:5 dilution of ultrasound gel (make sure you use the clear one, >> not the blue tinted stuff). >> > > > -- > ------------------------------------------------------------ > Steffen Dietzel, PD Dr. rer. nat > Ludwig-Maximilians-Universität München > Walter-Brendel-Zentrum für experimentelle Medizin (WBex) > Head of light microscopy > > Mail room: > Marchioninistr. 15, D-81377 München > > Building location: > Marchioninistr. 27, München-GroÃhadern This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research Ltd does not waive any rights if you have received this communication in error. The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research Ltd.  | 
			
 
	
	 | 
						Zac Arrac Atelaz | 
| 
					
		 | 
				
					
	 
				In reply to this post by Stephen Firth (Med)
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We have the Olympus FV10i with automatic water pump, it works fine, it keeps water continuosly on the lens tip in multipoint, time lapse, z-stack acquisitions with no trouble, even combining all of them, it also controls temperature and allows CO2 input working with a external perfussion pump in 35mm dish. I am not sure it can use something different to water but I also feel there is no need Gabriel OH > Date: Sat, 23 Jul 2011 08:25:40 +1000 > From: [hidden email] > Subject: Re: Alternative immersion media for a water immersion lens > To: [hidden email] > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > The other way to do it is a continual water source. > We have the Leica 63x with computer controlled water pump (no commercial > interest), but also use a lab-made system using a pipette, reservoir and > pump (designed by Steve Cody). > > We also use the Zeiss immersol and Cargille water RI oils but find they dry > over time and the quality of the image deteriorates. The water pump also > allow imaging multipoint experiments where there is considerable movement of > the objective. > Cheers, > Stephen > > On 23 July 2011 03:19, Farid Jalali <[hidden email]> wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Hello All. Thank you very much to everyone for your replies and insightful > > suggestions. It never ceases to amaze me what a wonderful community this > > is. > > Have a great day all. > > Cheers > > Farid > > > > On Thu, Jul 21, 2011 at 9:42 PM, Farid Jalali <[hidden email]> > > wrote: > > > > > Hello all, > > > Is anyone aware of a suitable alternative immersion media for a water > > > immersion lens? This lab is doing experiments at room temperature and the > > > scope is not enclosed therefore no humidity control. > > > Thanks very much all. > > > > > > Farid Jalali > > > > > > > > > -- > Stephen Firth > Manager of Advanced Optical Microscopy > Monash Micro Imaging > Monash University > Room G58 Building 75 > Phone 990 55612 > Mob 0400 695 425  | 
			
 
	
	 | 
						Lathrop, Kira L | 
| 
					
		 | 
				
					
	 
				In reply to this post by Farid Jalali
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Farid, You can use Genteal (Novartis Ophthalmics) RI 1.3. It's the media we use to couple a corneal confocal system to the patient's eye. It has a thick, gel-like consistency and so is easy to use on inverted scopes. The Cargille series AAA oil RI 1.335 works well but tends to run. Hope all is well, Kira  | 
			
 
	
	 | 
						Glen MacDonald-2 | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Kira, How long will Genteal last on a microscope before drying out or changing RI? regards, Glen Glen MacDonald Core for Communication Research Virginia Merrill Bloedel Hearing Research Center Box 357923 University of Washington Seattle, WA 98195-7923 USA (206) 616-4156 [hidden email] On Jul 25, 2011, at 4:40 AM, Lathrop, Kira L wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi Farid, > You can use Genteal (Novartis Ophthalmics) RI 1.3. It's the media we use to couple a corneal confocal system to the patient's eye. It has a thick, gel-like consistency and so is easy to use on inverted scopes. The Cargille series AAA oil RI 1.335 works well but tends to run. > Hope all is well, > Kira  | 
			
 
	
	 | 
						leoncio vergara | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I was wondering if there is any non toxic alternative help in segmentation of nucleus and cytosol compartments to measure protein translocation dynamics in 12-24 hrs time lapse experiments in live cultured cells. We are working with cell lines stably expressing both GFP and Strawberry fluorescent protein constructs. We want to follow the single cell translocation process and correlate the two signals over a period of 12-24 hrs. We are looking for a tool to aid in segmentation between both compartments for automated image analysis. The problem is that all the DNA binding dyes we have tried are cytotoxic. We have tried DRAQ5, DAPI, HOESCHT and several SITO dyes from molecular probes. We have not tried the Vybrant(r) DyeCycle stains from Invitrogen, violet and Ruby could be compatible with GFP and Strawberry, but a call to their technical support was not very encouraging. I am wondering if there are any long term live cell tracking dyes that can label the cytosol and give a "negative" image of the nucleus, be non toxic and be compatible with GFP and Strawberry. Invitrogen also has the CellLight reagents but they are base either on GFP or RFP so won't be a solution either since we are using those channels. I guess we could develop a far-red construct for triple FP labeling, but I was hoping for an easier solution... :) For additional reference: the microscope we are using is a Prairie Technologies Swept Field Confocal, we have 4 channels available (typical DAPI/FITC/TRITC/Cy5) to work. This is not an spectral imaging system. Thanks in advance, any suggestion would be greatly appreciated Leoncio Vergara Technical Director Optical Microscopy Core Galveston Texas  | 
			
 
	
	 | 
						David Baddeley | 
| 
					
		 | 
				
					
	
	 
		One option you could look at is using replication labellin
 
				***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Leoncio, One option you could look at is using replication labelling with a labelled nuclotide as available from e.g: http://www.jenabioscience.com/cms/en/1/catalog/1001 (no commercial interest - was just the first site I hit on in google. They seem to have conjugates covering the whole spectral range.) I'm not too sure on cytotoxicity, but know that cells labelled in this way can be observed through multiple divisions. The other option that comes to mind is fluorescent protein tagged histone, although you'd need to find something which worked spectrally. Both of these will give you a nuclear labelling rather than a cytoplasmic one. cheers, David ----- Original Message ---- From: "Vergara, Leoncio A." <[hidden email]> To: [hidden email] Sent: Tue, 26 July, 2011 9:20:01 AM Subject: Long term Nuclear labeling in live cells ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I was wondering if there is any non toxic alternative help in segmentation of nucleus and cytosol compartments to measure protein translocation dynamics in 12-24 hrs time lapse experiments in live cultured cells. We are working with cell lines stably expressing both GFP and Strawberry fluorescent protein constructs. We want to follow the single cell translocation process and correlate the two signals over a period of 12-24 hrs. We are looking for a tool to aid in segmentation between both compartments for automated image analysis. The problem is that all the DNA binding dyes we have tried are cytotoxic. We have tried DRAQ5, DAPI, HOESCHT and several SITO dyes from molecular probes. We have not tried the Vybrant(r) DyeCycle stains from Invitrogen, violet and Ruby could be compatible with GFP and Strawberry, but a call to their technical support was not very encouraging. I am wondering if there are any long term live cell tracking dyes that can label the cytosol and give a "negative" image of the nucleus, be non toxic and be compatible with GFP and Strawberry. Invitrogen also has the CellLight reagents but they are base either on GFP or RFP so won't be a solution either since we are using those channels. I guess we could develop a far-red construct for triple FP labeling, but I was hoping for an easier solution... :) For additional reference: the microscope we are using is a Prairie Technologies Swept Field Confocal, we have 4 channels available (typical DAPI/FITC/TRITC/Cy5) to work. This is not an spectral imaging system. Thanks in advance, any suggestion would be greatly appreciated Leoncio Vergara Technical Director Optical Microscopy Core Galveston Texas  | 
			
 
	
	 | 
						Paul Rigby-2 | 
| 
					
		 | 
				
					
	 
				In reply to this post by leoncio vergara
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Leoncio, While I can't suggest a solution, I am interested in your comments on the dyes you have tried for your live cell imaging, specifically, the UV/violet excited dyes Hoechst and DAPI. My experience has been that Hoechst 33342 will stain nuclei in living cells and has not proved toxic over 24 hours of imaging, albeit using very low laser powers and only infrequent imaging. Were you using Hoechst 33342 or Hoechst 33258 (or the slightly longer wavelength excited Hoechst 34580)? Also, what concentration were you using? Frigault et al (J Cell Sci, 122, 753-767, 2009) suggest that to avoid toxicity effects in live cell imaging, these dyes may need to be used at 10-100x more dilute concentrations than usually recommended. Also you mentioned using DAPI for nuclear staining. This probe is relatively live cell membrane impermeant and usually requires quite high concentrations to get significant nuclear labelling in living cells. At high concentrations DAPI is also toxic so I am not surprised you had problems with this dye. As an aside, has anyone seen inhibition of cell division when using these DNA intercalating dyes? Some people suggest that cell division is inhibited, but others have not reported any problems. What is the concensus? Regards Paul Assoc. Prof. Paul Rigby Centre for Microscopy, Characterisation & Analysis (M510) The University of Western Australia 35 Stirling Highway Crawley WA 6007 Australia -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Vergara, Leoncio A. Sent: Tuesday, 26 July 2011 5:20 AM To: [hidden email] Subject: Long term Nuclear labeling in live cells ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I was wondering if there is any non toxic alternative help in segmentation of nucleus and cytosol compartments to measure protein translocation dynamics in 12-24 hrs time lapse experiments in live cultured cells. We are working with cell lines stably expressing both GFP and Strawberry fluorescent protein constructs. We want to follow the single cell translocation process and correlate the two signals over a period of 12-24 hrs. We are looking for a tool to aid in segmentation between both compartments for automated image analysis. The problem is that all the DNA binding dyes we have tried are cytotoxic. We have tried DRAQ5, DAPI, HOESCHT and several SITO dyes from molecular probes. We have not tried the Vybrant(r) DyeCycle stains from Invitrogen, violet and Ruby could be compatible with GFP and Strawberry, but a call to their technical support was not very encouraging. I am wondering if there are any long term live cell tracking dyes that can label the cytosol and give a "negative" image of the nucleus, be non toxic and be compatible with GFP and Strawberry. Invitrogen also has the CellLight reagents but they are base either on GFP or RFP so won't be a solution either since we are using those channels. I guess we could develop a far-red construct for triple FP labeling, but I was hoping for an easier solution... :) For additional reference: the microscope we are using is a Prairie Technologies Swept Field Confocal, we have 4 channels available (typical DAPI/FITC/TRITC/Cy5) to work. This is not an spectral imaging system. Thanks in advance, any suggestion would be greatly appreciated Leoncio Vergara Technical Director Optical Microscopy Core Galveston Texas  | 
			
 
	
	 | 
						Cameron Nowell | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Paul, I have used Hoechst 33342 in the past for live imaging both primary and cultured cells. After about 24 hours they will dye (usually quite spectacularly). But i have managed to image cells divinging with Hoechst, but the best is only ever one round of division. Cheers Cam -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Paul Rigby Sent: Tuesday, 26 July 2011 11:46 AM To: [hidden email] Subject: Re: Long term Nuclear labeling in live cells ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Leoncio, While I can't suggest a solution, I am interested in your comments on the dyes you have tried for your live cell imaging, specifically, the UV/violet excited dyes Hoechst and DAPI. My experience has been that Hoechst 33342 will stain nuclei in living cells and has not proved toxic over 24 hours of imaging, albeit using very low laser powers and only infrequent imaging. Were you using Hoechst 33342 or Hoechst 33258 (or the slightly longer wavelength excited Hoechst 34580)? Also, what concentration were you using? Frigault et al (J Cell Sci, 122, 753-767, 2009) suggest that to avoid toxicity effects in live cell imaging, these dyes may need to be used at 10-100x more dilute concentrations than usually recommended. Also you mentioned using DAPI for nuclear staining. This probe is relatively live cell membrane impermeant and usually requires quite high concentrations to get significant nuclear labelling in living cells. At high concentrations DAPI is also toxic so I am not surprised you had problems with this dye. As an aside, has anyone seen inhibition of cell division when using these DNA intercalating dyes? Some people suggest that cell division is inhibited, but others have not reported any problems. What is the concensus? Regards Paul Assoc. Prof. Paul Rigby Centre for Microscopy, Characterisation & Analysis (M510) The University of Western Australia 35 Stirling Highway Crawley WA 6007 Australia -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Vergara, Leoncio A. Sent: Tuesday, 26 July 2011 5:20 AM To: [hidden email] Subject: Long term Nuclear labeling in live cells ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I was wondering if there is any non toxic alternative help in segmentation of nucleus and cytosol compartments to measure protein translocation dynamics in 12-24 hrs time lapse experiments in live cultured cells. We are working with cell lines stably expressing both GFP and Strawberry fluorescent protein constructs. We want to follow the single cell translocation process and correlate the two signals over a period of 12-24 hrs. We are looking for a tool to aid in segmentation between both compartments for automated image analysis. The problem is that all the DNA binding dyes we have tried are cytotoxic. We have tried DRAQ5, DAPI, HOESCHT and several SITO dyes from molecular probes. We have not tried the Vybrant(r) DyeCycle stains from Invitrogen, violet and Ruby could be compatible with GFP and Strawberry, but a call to their technical support was not very encouraging. I am wondering if there are any long term live cell tracking dyes that can label the cytosol and give a "negative" image of the nucleus, be non toxic and be compatible with GFP and Strawberry. Invitrogen also has the CellLight reagents but they are base either on GFP or RFP so won't be a solution either since we are using those channels. I guess we could develop a far-red construct for triple FP labeling, but I was hoping for an easier solution... :) For additional reference: the microscope we are using is a Prairie Technologies Swept Field Confocal, we have 4 channels available (typical DAPI/FITC/TRITC/Cy5) to work. This is not an spectral imaging system. Thanks in advance, any suggestion would be greatly appreciated Leoncio Vergara Technical Director Optical Microscopy Core Galveston Texas This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research Ltd does not waive any rights if you have received this communication in error. The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research Ltd.  | 
			
 
	
	 | 
						George McNamara | 
| 
					
		 | 
				
					
	 
				In reply to this post by Paul Rigby-2
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Paul, Hoechst 33### are minor groove binders, not intercalators (at least at usual concentrations). With respect to Frigault et al review - they pan Leibowitz media. Not all mammalian cells behave like CHO-K1 (their Fig 1). One of my users found Leibowitz works well for her lymphoma cells (and possibly normal B-cells). I recommend not taking reviews as gospel or any single cell line as representative for all mammalian (or other) cells. Enjoy, George On 7/25/2011 9:46 PM, Paul Rigby wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi Leoncio, > While I can't suggest a solution, I am interested in your comments on the dyes you have tried for your live cell imaging, specifically, the UV/violet excited dyes Hoechst and DAPI. > > My experience has been that Hoechst 33342 will stain nuclei in living cells and has not proved toxic over 24 hours of imaging, albeit using very low laser powers and only infrequent imaging. Were you using Hoechst 33342 or Hoechst 33258 (or the slightly longer wavelength excited Hoechst 34580)? Also, what concentration were you using? Frigault et al (J Cell Sci, 122, 753-767, 2009) suggest that to avoid toxicity effects in live cell imaging, these dyes may need to be used at 10-100x more dilute concentrations than usually recommended. > > Also you mentioned using DAPI for nuclear staining. This probe is relatively live cell membrane impermeant and usually requires quite high concentrations to get significant nuclear labelling in living cells. At high concentrations DAPI is also toxic so I am not surprised you had problems with this dye. > > As an aside, has anyone seen inhibition of cell division when using these > DNA intercalating dyes? > Some people suggest that cell division is inhibited, but others have not reported any problems. What is the concensus? > > Regards > Paul > > Assoc. Prof. Paul Rigby > Centre for Microscopy, Characterisation& Analysis (M510) > The University of Western Australia > 35 Stirling Highway > Crawley WA 6007 > Australia > > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Vergara, Leoncio A. > Sent: Tuesday, 26 July 2011 5:20 AM > To: [hidden email] > Subject: Long term Nuclear labeling in live cells > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I was wondering if there is any non toxic alternative help in segmentation of nucleus and cytosol compartments to measure protein translocation dynamics in 12-24 hrs time lapse experiments in live cultured cells. > > We are working with cell lines stably expressing both GFP and Strawberry fluorescent protein constructs. We want to follow the single cell translocation process and correlate the two signals over a period of 12-24 hrs. We are looking for a tool to aid in segmentation between both compartments for automated image analysis. The problem is that all the DNA binding dyes we have tried are cytotoxic. We have tried DRAQ5, DAPI, HOESCHT and several SITO dyes from molecular probes. We have not tried the Vybrant(r) DyeCycle stains from Invitrogen, violet and Ruby could be compatible with GFP and Strawberry, but a call to their technical support was not very encouraging. > > I am wondering if there are any long term live cell tracking dyes that can label the cytosol and give a "negative" image of the nucleus, be non toxic and be compatible with GFP and Strawberry. > > Invitrogen also has the CellLight reagents but they are base either on GFP or RFP so won't be a solution either since we are using those channels. I guess we could develop a far-red construct for triple FP labeling, but I was hoping for an easier solution... :) > > For additional reference: the microscope we are using is a Prairie Technologies Swept Field Confocal, we have 4 channels available (typical DAPI/FITC/TRITC/Cy5) to work. This is not an spectral imaging system. > > Thanks in advance, any suggestion would be greatly appreciated > > Leoncio Vergara > Technical Director > Optical Microscopy Core > Galveston > Texas > > -- George McNamara, PhD Analytical Imaging Core Facility University of Miami  | 
			
 
	
	 | 
						Claudia Florindo | 
| 
					
		 | 
				
					
	 
				In reply to this post by leoncio vergara
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Leoncio Dapi is probably not good because you use UV light for imaging, and cells do dye under UVs. I have imaged cells for 24h (time points every 3 min) labeled with GFP-Histone 2B. If you can get histone 2B with a in fusion with a fluorescent protein that goes to the CY5 channel it should work. You could also try Histone GFP and do a bright field /DIC image where you can clearly see the nucleus and the cytoplasm. Or DIC alone merged with the other 2 florescent channels images that you have can also help you to distinguish between the nucleus and the cytoplasm. Good luck Claudia. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Vergara, Leoncio A. Sent: segunda-feira, 25 de Julho de 2011 22:20 To: [hidden email] Subject: Long term Nuclear labeling in live cells ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I was wondering if there is any non toxic alternative help in segmentation of nucleus and cytosol compartments to measure protein translocation dynamics in 12-24 hrs time lapse experiments in live cultured cells. We are working with cell lines stably expressing both GFP and Strawberry fluorescent protein constructs. We want to follow the single cell translocation process and correlate the two signals over a period of 12-24 hrs. We are looking for a tool to aid in segmentation between both compartments for automated image analysis. The problem is that all the DNA binding dyes we have tried are cytotoxic. We have tried DRAQ5, DAPI, HOESCHT and several SITO dyes from molecular probes. We have not tried the Vybrant(r) DyeCycle stains from Invitrogen, violet and Ruby could be compatible with GFP and Strawberry, but a call to their technical support was not very encouraging. I am wondering if there are any long term live cell tracking dyes that can label the cytosol and give a "negative" image of the nucleus, be non toxic and be compatible with GFP and Strawberry. Invitrogen also has the CellLight reagents but they are base either on GFP or RFP so won't be a solution either since we are using those channels. I guess we could develop a far-red construct for triple FP labeling, but I was hoping for an easier solution... :) For additional reference: the microscope we are using is a Prairie Technologies Swept Field Confocal, we have 4 channels available (typical DAPI/FITC/TRITC/Cy5) to work. This is not an spectral imaging system. Thanks in advance, any suggestion would be greatly appreciated Leoncio Vergara Technical Director Optical Microscopy Core Galveston Texas ----- Nenhum vírus encontrado nessa mensagem. Verificado por AVG - www.avgbrasil.com.br Versão: 10.0.1390 / Banco de dados de vírus: 1518/3788 - Data de Lançamento: 07/25/11  | 
			
 
	
	 | 
						Confocal2011 | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** While our work is with budding yeast and I have virtually no experience outside of yeast, the fact that a fluorescent protein tag was mentioned by someone else suggests that maybe it would work for you. In that case I will just add that we use histone 2B with a C-terminal mCherry tag with amazing results. On Tue, Jul 26, 2011 at 3:53 AM, Claudia Florindo <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi Leoncio > > Dapi is probably not good because you use UV light for imaging, and cells > do > dye under UVs. > > I have imaged cells for 24h (time points every 3 min) labeled with > GFP-Histone 2B. If you can get histone 2B with a in fusion with a > fluorescent protein that goes to the CY5 channel it should work. > > You could also try Histone GFP and do a bright field /DIC image where you > can clearly see the nucleus and the cytoplasm. Or DIC alone merged with the > other 2 florescent channels images that you have can also help you to > distinguish between the nucleus and the cytoplasm. > > Good luck > > Claudia. > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On > Behalf Of Vergara, Leoncio A. > Sent: segunda-feira, 25 de Julho de 2011 22:20 > To: [hidden email] > Subject: Long term Nuclear labeling in live cells > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I was wondering if there is any non toxic alternative help in segmentation > of nucleus and cytosol compartments to measure protein translocation > dynamics in 12-24 hrs time lapse experiments in live cultured cells. > > We are working with cell lines stably expressing both GFP and Strawberry > fluorescent protein constructs. We want to follow the single cell > translocation process and correlate the two signals over a period of 12-24 > hrs. We are looking for a tool to aid in segmentation between both > compartments for automated image analysis. The problem is that all the DNA > binding dyes we have tried are cytotoxic. We have tried DRAQ5, DAPI, > HOESCHT > and several SITO dyes from molecular probes. We have not tried the > Vybrant(r) DyeCycle stains from Invitrogen, violet and Ruby could be > compatible with GFP and Strawberry, but a call to their technical support > was not very encouraging. > > I am wondering if there are any long term live cell tracking dyes that can > label the cytosol and give a "negative" image of the nucleus, be non toxic > and be compatible with GFP and Strawberry. > > Invitrogen also has the CellLight reagents but they are base either on GFP > or RFP so won't be a solution either since we are using those channels. I > guess we could develop a far-red construct for triple FP labeling, but I > was > hoping for an easier solution... :) > > For additional reference: the microscope we are using is a Prairie > Technologies Swept Field Confocal, we have 4 channels available (typical > DAPI/FITC/TRITC/Cy5) to work. This is not an spectral imaging system. > > Thanks in advance, any suggestion would be greatly appreciated > > Leoncio Vergara > Technical Director > Optical Microscopy Core > Galveston > Texas > > ----- > Nenhum vírus encontrado nessa mensagem. > Verificado por AVG - www.avgbrasil.com.br > Versão: 10.0.1390 / Banco de dados de vírus: 1518/3788 - Data de > Lançamento: > 07/25/11 >  | 
			
 
	
	 | 
						Masur, Sandra | 
| 
					
		 | 
				
					
	 
				In reply to this post by Cameron Nowell
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear all who indicate that various nuclear dyes will cause > cultured cells....to... dye (usually quite > spectacularly) You mean that the "dye" (addition of a color) will cause the cells to "die" (cease living). English spelling can be very confusing! On Jul 25, 2011, at 10:24 PM, Cameron Nowell wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi Paul, > > I have used Hoechst 33342 in the past for live imaging both primary and > cultured cells. After about 24 hours they will dye (usually quite > spectacularly). But i have managed to image cells divinging with > Hoechst, but the best is only ever one round of division. > > Cheers > > Cam > > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Paul Rigby > Sent: Tuesday, 26 July 2011 11:46 AM > To: [hidden email] > Subject: Re: Long term Nuclear labeling in live cells > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi Leoncio, > While I can't suggest a solution, I am interested in your comments on > the dyes you have tried for your live cell imaging, specifically, the > UV/violet excited dyes Hoechst and DAPI. > > My experience has been that Hoechst 33342 will stain nuclei in living > cells and has not proved toxic over 24 hours of imaging, albeit using > very low laser powers and only infrequent imaging. Were you using > Hoechst 33342 or Hoechst 33258 (or the slightly longer wavelength > excited Hoechst 34580)? Also, what concentration were you using? > Frigault et al (J Cell Sci, 122, 753-767, 2009) suggest that to avoid > toxicity effects in live cell imaging, these dyes may need to be used at > 10-100x more dilute concentrations than usually recommended. > > Also you mentioned using DAPI for nuclear staining. This probe is > relatively live cell membrane impermeant and usually requires quite high > concentrations to get significant nuclear labelling in living cells. At > high concentrations DAPI is also toxic so I am not surprised you had > problems with this dye. > > As an aside, has anyone seen inhibition of cell division when using > these DNA intercalating dyes? Some people suggest that cell division is > inhibited, but others have not reported any problems. What is the > concensus? > > Regards > Paul > > Assoc. Prof. Paul Rigby > Centre for Microscopy, Characterisation & Analysis (M510) > The University of Western Australia > 35 Stirling Highway > Crawley WA 6007 > Australia > > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Vergara, Leoncio A. > Sent: Tuesday, 26 July 2011 5:20 AM > To: [hidden email] > Subject: Long term Nuclear labeling in live cells > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I was wondering if there is any non toxic alternative help in > segmentation of nucleus and cytosol compartments to measure protein > translocation dynamics in 12-24 hrs time lapse experiments in live > cultured cells. > > We are working with cell lines stably expressing both GFP and Strawberry > fluorescent protein constructs. We want to follow the single cell > translocation process and correlate the two signals over a period of > 12-24 hrs. We are looking for a tool to aid in segmentation between both > compartments for automated image analysis. The problem is that all the > DNA binding dyes we have tried are cytotoxic. We have tried DRAQ5, DAPI, > HOESCHT and several SITO dyes from molecular probes. We have not tried > the Vybrant(r) DyeCycle stains from Invitrogen, violet and Ruby could be > compatible with GFP and Strawberry, but a call to their technical > support was not very encouraging. > > I am wondering if there are any long term live cell tracking dyes that > can label the cytosol and give a "negative" image of the nucleus, be non > toxic and be compatible with GFP and Strawberry. > > Invitrogen also has the CellLight reagents but they are base either on > GFP or RFP so won't be a solution either since we are using those > channels. I guess we could develop a far-red construct for triple FP > labeling, but I was hoping for an easier solution... :) > > For additional reference: the microscope we are using is a Prairie > Technologies Swept Field Confocal, we have 4 channels available (typical > DAPI/FITC/TRITC/Cy5) to work. This is not an spectral imaging system. > > Thanks in advance, any suggestion would be greatly appreciated > > Leoncio Vergara > Technical Director > Optical Microscopy Core > Galveston > Texas > > > This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research Ltd does not waive any rights if you have received this communication in error. > The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research Ltd.  | 
			