Csúcs Gábor-3 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear All, A user of us uses Oxyrase in her experiments to reduce the amount of bleaching (tracking YFP labelled spots). Though we are working on optimizing her experiments also in terms of optics, still the question came up: what alternatives do exist to Oxyrase and what are your experiences with those? Clearly we would need something that works with living cells. By looking on the Internet I have seen the special imaging biffer of Evrogen. Is anyone using it? Or alternative suggestions? Thanks Gabor Gabor Csucs ScopeM Wolgang-Pauli Str. 14 CH-8093 Zurich Phone: +41 44 633 6221 |
|
Steffen Dietzel |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** A former colleague used Trolox and was happy about it. It's a water-solubale Vitamin E derivate. Steffen Am 23.01.2015 um 12:03 schrieb Csúcs Gábor: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear All, > > A user of us uses Oxyrase in her experiments to reduce the amount of > bleaching (tracking YFP labelled spots). Though we are working on > optimizing her experiments also in terms of optics, still the question > came up: what alternatives do exist to Oxyrase and what are your > experiences with those? Clearly we would need something that works with > living cells. By looking on the Internet I have seen the special imaging > biffer of Evrogen. Is anyone using it? Or alternative suggestions? > > > Thanks Gabor > > Gabor Csucs > ScopeM > Wolgang-Pauli Str. 14 > CH-8093 Zurich > > Phone: +41 44 633 6221 > -- ------------------------------------------------------------ Steffen Dietzel, PD Dr. rer. nat Ludwig-Maximilians-Universität München Walter-Brendel-Zentrum für experimentelle Medizin (WBex) Head of light microscopy Marchioninistr. 27 D-81377 München Phone: +49/89/2180-76509 Fax-to-email: +49/89/2180-9976509 skype: steffendietzel e-mail: [hidden email] -- ------------------------------------------------------------ Steffen Dietzel, PD Dr. rer. nat Ludwig-Maximilians-Universität München Walter-Brendel-Zentrum für experimentelle Medizin (WBex) Head of light microscopy Marchioninistr. 27 D-81377 München Germany |
|
Hallworth, Richard J. |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Steffen, Gabor - I have never heard of Trolox or any other such agent working for fluorescent proteins, though they work brilliantly for organic fluorophores. The bleaching mechanisms for fps are unknown, as far as I know. Anyone know different? Rick Richard Hallworth, Ph.D., Professor Director, Integrated Biomedical Imaging Facility Department of Biomedical Sciences Creighton University School of Medicine 2500 California Plaza Omaha NE 68178 Ph: (402) 280-3057 FAX: (402) 280-2690 Email: [hidden email] http://biomedsci.creighton.edu/hallworth -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Steffen Dietzel Sent: Friday, January 23, 2015 12:03 PM To: [hidden email] Subject: Re: Anti-bleaching agents in live-cell microscopy ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** A former colleague used Trolox and was happy about it. It's a water-solubale Vitamin E derivate. Steffen Am 23.01.2015 um 12:03 schrieb Csúcs Gábor: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear All, > > A user of us uses Oxyrase in her experiments to reduce the amount of > bleaching (tracking YFP labelled spots). Though we are working on > optimizing her experiments also in terms of optics, still the question > came up: what alternatives do exist to Oxyrase and what are your > experiences with those? Clearly we would need something that works > with living cells. By looking on the Internet I have seen the special > imaging biffer of Evrogen. Is anyone using it? Or alternative suggestions? > > > Thanks Gabor > > Gabor Csucs > ScopeM > Wolgang-Pauli Str. 14 > CH-8093 Zurich > > Phone: +41 44 633 6221 > -- ------------------------------------------------------------ Steffen Dietzel, PD Dr. rer. nat Ludwig-Maximilians-Universität München Walter-Brendel-Zentrum für experimentelle Medizin (WBex) Head of light microscopy Marchioninistr. 27 D-81377 München Phone: +49/89/2180-76509 Fax-to-email: +49/89/2180-9976509 skype: steffendietzel e-mail: [hidden email] -- ------------------------------------------------------------ Steffen Dietzel, PD Dr. rer. nat Ludwig-Maximilians-Universität München Walter-Brendel-Zentrum für experimentelle Medizin (WBex) Head of light microscopy Marchioninistr. 27 D-81377 München Germany |
|
WHEELER Ann |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Gabor I've tried Oxyrase too, it does work (by black magic) and I definitely would never use it with any leukocytes. I have had some success with Ascorbic acid as well, although thr most ideal solution would be to move to an FP with a higher quantum yeild such as Clover or mNeon green, but cloning is a pain. Ann Dr Ann Wheeler Head of Advanced Imaging / ESRIC@IGMM MRC IGMM Crewe Road South University of Edinburgh Edinburgh EH4 2XU UK ________________________________________ From: Confocal Microscopy List <[hidden email]> on behalf of Hallworth, Richard J. <[hidden email]> Sent: 23 January 2015 21:16 To: [hidden email] Subject: Re: Anti-bleaching agents in live-cell microscopy ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Steffen, Gabor - I have never heard of Trolox or any other such agent working for fluorescent proteins, though they work brilliantly for organic fluorophores. The bleaching mechanisms for fps are unknown, as far as I know. Anyone know different? Rick Richard Hallworth, Ph.D., Professor Director, Integrated Biomedical Imaging Facility Department of Biomedical Sciences Creighton University School of Medicine 2500 California Plaza Omaha NE 68178 Ph: (402) 280-3057 FAX: (402) 280-2690 Email: [hidden email] http://biomedsci.creighton.edu/hallworth -- The University of Edinburgh is a charitable body, registered in Scotland, with registration number SC005336. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Steffen Dietzel Sent: Friday, January 23, 2015 12:03 PM To: [hidden email] Subject: Re: Anti-bleaching agents in live-cell microscopy ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** A former colleague used Trolox and was happy about it. It's a water-solubale Vitamin E derivate. Steffen Am 23.01.2015 um 12:03 schrieb Csúcs Gábor: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear All, > > A user of us uses Oxyrase in her experiments to reduce the amount of > bleaching (tracking YFP labelled spots). Though we are working on > optimizing her experiments also in terms of optics, still the question > came up: what alternatives do exist to Oxyrase and what are your > experiences with those? Clearly we would need something that works > with living cells. By looking on the Internet I have seen the special > imaging biffer of Evrogen. Is anyone using it? Or alternative suggestions? > > > Thanks Gabor > > Gabor Csucs > ScopeM > Wolgang-Pauli Str. 14 > CH-8093 Zurich > > Phone: +41 44 633 6221 > -- ------------------------------------------------------------ Steffen Dietzel, PD Dr. rer. nat Ludwig-Maximilians-Universität München Walter-Brendel-Zentrum für experimentelle Medizin (WBex) Head of light microscopy Marchioninistr. 27 D-81377 München Phone: +49/89/2180-76509 Fax-to-email: +49/89/2180-9976509 skype: steffendietzel e-mail: [hidden email] -- ------------------------------------------------------------ Steffen Dietzel, PD Dr. rer. nat Ludwig-Maximilians-Universität München Walter-Brendel-Zentrum für experimentelle Medizin (WBex) Head of light microscopy Marchioninistr. 27 D-81377 München Germany |
|
George McNamara |
|
|
In reply to this post by Csúcs Gábor-3
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Gabor, Evrogen DMEMgfp2 ... note option to add Rutin (also available from sigma-aldrich) on last page of the pdf. http://www.evrogen.com/medium_DMEM_gfp-description/DMEMgfp2.pdf Marker Gene Technologies Opti-Klear Live Cell Imaging Buffer http://www.markergene.com/ProductDetails.php/M1919 http://www.markergene.com/ProductDetails.php/M1898 since you are in Europe, probably easier for you to get Evrogen's. For researchers in the US, Evrogen has distributors here. Some suggestions for everyone: * EGFP is soooo 1996 ... are you using a 1996 PC, cell phone or car? ... middle of http://works.bepress.com/gmcnamara/63/ has a list of state of the art color FPs. See also Adam 2014 paper (ref below) and Kurt Thorn's FP properties page, http://nic.ucsf.edu/FPvisualization/ * multimerize! ... Steven Vogel (Nguyen 2012) has published V6 and even tastier and more powerful engine, unpublished V8. * "go binary" - see ://works.bepress.com/gmcnamara/65 ... see also Burgess 2012 (PCNA Chromobodies) and Stasevich 2014 (FabLEM; can use alternative ways to get into cells) * don't overexpress! Just trashes your SNR and produces O2 radicals all over your cells - see "65" FingR references for one way, or TetOn/Off (though many tetracyclines are fluorescent), Cumate, RheoSwitch, etc ligand controllable transcription factors. Adam V. Phototransformable fluorescent proteins: which one for which application? Histochem Cell Biol. 2014 Jul;142(1):19-41. doi:10.1007/s00418-014-1190-5. PubMed PMID: 24522394. Nguyen TA, Sarkar P, Veetil JV, Koushik SV, Vogel SS. Fluorescence polarization and fluctuation analysis monitors subunit proximity, stoichiometry, and protein complex hydrodynamics. PLoS One. 2012;7(5):e38209. doi: 10.1371/journal.pone.0038209. Epub 2012 May 30. PubMed PMID: 22666486; PubMed Central PMCID: PMC3364239. There are also nanobodies/chromobodies to detect or amplify FPs, see http://www.allelebiotech.com/nab and the Burgess paper Burgess A, Lorca T, Castro A. Quantitative live imaging of endogenous DNA replication in mammalian cells. PLoS One. 2012;7(9):e45726. doi: 10.1371/journal.pone.0045726. Epub 2012 Sep 20. PubMed PMID: 23029203; PubMed Central PMCID: PMC3447815. and Fab-dye equivalent Stasevich TJ, Hayashi-Takanaka Y, Sato Y, Maehara K, Ohkawa Y, Sakata-Sogawa K, Tokunaga M, Nagase T, Nozaki N, McNally JG, Kimura H. Regulation of RNA polymerase II activation by histone acetylation in single living cells. Nature. 2014 Dec 11;516(7530):272-5. doi: 10.1038/nature13714. Epub 2014 Sep 21. PubMed PMID: 25252976. Happy IYL2015 http://www.light2015.org/Home.html George On 1/23/2015 5:03 AM, Csúcs Gábor wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear All, > > A user of us uses Oxyrase in her experiments to reduce the amount of > bleaching (tracking YFP labelled spots). Though we are working on > optimizing her experiments also in terms of optics, still the question > came up: what alternatives do exist to Oxyrase and what are your > experiences with those? Clearly we would need something that works with > living cells. By looking on the Internet I have seen the special imaging > biffer of Evrogen. Is anyone using it? Or alternative suggestions? > > > Thanks Gabor > > Gabor Csucs > ScopeM > Wolgang-Pauli Str. 14 > CH-8093 Zurich > > Phone: +41 44 633 6221 > > -- George McNamara, Ph.D. Single Cells Analyst L.J.N. Cooper Lab University of Texas M.D. Anderson Cancer Center Houston, TX 77054 Tattletales http://works.bepress.com/gmcnamara/42 |
|
Cole, Richard W (HEALTH) |
|
|
In reply to this post by Csúcs Gábor-3
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Just to throw another option out there: LiveLight http://www.cellgs.com/Shop/Cell-Culture-Systems/LiveLight.html Rich Richard Cole Research Scientist V Director: Advanced Light Microscopy & Image Analysis Core Wadsworth Center Research Assistant Professor Dept. of Biomedical Sciences School of Public Health State University of New York P.O. Box 509 Albany N.Y. 12201-0509 518-474-7048 Phone 518-473-2895 Fax Website www.wadsworth.org/cores/alm/index.htm From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of CONFOCALMICROSCOPY automatic digest system Sent: Monday, January 26, 2015 1:04 AM To: [hidden email] Subject: CONFOCALMICROSCOPY Digest - 23 Jan 2015 to 25 Jan 2015 (#2015-22) CONFOCALMICROSCOPY Digest - 23 Jan 2015 to 25 Jan 2015 (#2015-22) Table of contents: . Anti-bleaching agents in live-cell microscopy 1. Anti-bleaching agents in live-cell microscopy o Re: Anti-bleaching agents in live-cell microscopy (01/25) From: George McNamara <[hidden email]> Browse the CONFOCALMICROSCOPY online archives. |
|
Vitaly Boyko |
|
|
In reply to this post by George McNamara
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Junior Digital Microscopy Technician Opening at Memorial Sloan-Kettering Cancer Center’s Molecular Cytology Core Facility Molecular Cytology Core Facility (MCCF) at the Memorial Sloan-Kettering Cancer Center is seeking an entry-level Junior Digital Microscopy Technician to support our daily operation. The MCCF is equipped with state of the art optical imaging systems, including Leica SP5 and SP8 confocal microscopes, a spinning disk and a Zeiss line-scanning confocals, and several widefield microscopes, digital slide scanners and also an Atomic Force Microscope. A number of widefield and confocal systems are equipped for live imaging. The image analysis software used includes Metamorph, Amira, Imaris, AutoDeblur, FIJI, MATLAB, Voloom. Responsibilities:· Train and assist users with imaging equipment· Assist in post-acquisition processing and analysis of the data, writing macros and scripts for image acquisition, processing and analysis· Maintain, calibrate, and monitor the performance of imaging systems· Stay informed with the hardware and software advances in optical microscopy· Work in collaborative settingsQualifications and Skills:· Bachelor’s degree in Biology or a related field with very strong computational background· Experience with advanced fluorescent microscopy is preferred, but training will be provided· Prior experience with image analysis would be significant plus· Excellent communication skills and ability to work in a team environment are very important· Strong organizational skills, flexibility in working hours and ability to multi-task· Interest in biomedical research and working with students, post-doctoral and clinical fellows, and faculty performing research requiring generation of 'high end' and state- of-the-art light microscopy imaging dataMemorial Sloan-Kettering Cancer Center is one of the top ranked leading cancer research and teaching centers, located in Upper East Side of Manhattan. We offer a competitive salary, comprehensive benefits and an excellent working environment. Please e-mail your CV and cover letter to Madeline Leung, [hidden email] Yevgeniy RominDigital Microscopy SpecialistMolecular Cytology Core FacilityMemorial Sloan-Kettering Cancer Center415 East 68th St. ZRC-1834New York, NY 10065(646) 888-2186, x2186 |