Login  Register

Anyone uses UPS units to protect your confocal/2-P systems?

classic Classic list List threaded Threaded
12 messages Options Options
Embed post
Permalink
Mario Methot Mario Methot
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Anyone uses UPS units to protect your confocal/2-P systems?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi all,

We are unfortunately facing many unplanned (random) electricity outages/voltage variations in our Institute (it's a very old building) and our systems (and particularly lasers) don't like it too much... I was wondering if some of you are using UPS units to protect your confocal/2-P systems? If so what type of UPS are you using? All the info you could provide would be most welcome, thanks in advance.

Mario
Colin Haig-2 Colin Haig-2
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Anyone uses UPS units to protect your confocal/2-P systems?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Mario,
We use APC Back UPS Pro sinewave units for a lot of sensitive equipment including CCD and CMOS cameras, test equipment, and other science-related instruments.
They have an excellent track record and reliability.
Eg BR1500VA and other models.
https://www.apc.com
Good luck.
Colin

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Mario Methot
Sent: February 22, 2021 8:50 AM
To: [hidden email]
Subject: Anyone uses UPS units to protect your confocal/2-P systems?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi all,

We are unfortunately facing many unplanned (random) electricity outages/voltage variations in our Institute (it's a very old building) and our systems (and particularly lasers) don't like it too much... I was wondering if some of you are using UPS units to protect your confocal/2-P systems? If so what type of UPS are you using? All the info you could provide would be most welcome, thanks in advance.

Mario
Sylvie Le Guyader Sylvie Le Guyader
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Anyone uses UPS units to protect your confocal/2-P systems?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

When we designed our new lab, we investigated which equipment is subject to damage.
Our multiphoton laser (Coherent chameleon Ultra II) has a built in safeguard with an internal battery sufficient to cool down the laser. The software reports the state of the battery. This is not written in the laser specs so it is worth asking the company if they have such a system or not.
Argon lasers will get their lifetime reduced if repeatedly turned off without cooling but there is no impact if it happens seldom. This is a problem if it happens often as seems to be the case in your institute.
The rest of the equipment (cameras, computers...) can be turned off suddenly without damage.

However the main problem is the spike that occurs when one turns the power back on. This is very damaging for a lot of equipment including the computers. It is very important to have a good spike protection system. Alternatively unplug all equipment before the current returns or setting the fuse back on and painfully replug them a few at a time.

Med vänlig hälsning / Best regards

Sylvie

@@@@@@@@@@@@@@@@@@@@@@@@
Sylvie Le Guyader, PhD
Live Cell Imaging Facility Manager
Karolinska Institutet- Bionut Dpt
Blickagången 16,
Room 7362 (lab)/7840 (office)
14157 Huddinge, Sweden
mobile: +46 (0) 73 733 5008
LCI website
Follow our microscopy blog!

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Colin Haig
Sent: 22 February 2021 15:09
To: [hidden email]
Subject: Re: Anyone uses UPS units to protect your confocal/2-P systems?

*****
To join, leave or search the confocal microscopy listserv, go to:
https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&amp;data=04%7C01%7Csylvie.le.guyader%40KI.SE%7C10fea0738af64c0d5d3c08d8d73b6687%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637495997440954136%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=xwnwC4w68SIHjqvlBLSjolIQ7YU6dg6kGNhV4JUNyYk%3D&amp;reserved=0
Post images on https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&amp;data=04%7C01%7Csylvie.le.guyader%40KI.SE%7C10fea0738af64c0d5d3c08d8d73b6687%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637495997440954136%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=65UMEP7SwGjqDd9vSSAUsNWAnLTR9gae5cmqrzMrkVM%3D&amp;reserved=0 and include the link in your posting.
*****

Hi Mario,
We use APC Back UPS Pro sinewave units for a lot of sensitive equipment including CCD and CMOS cameras, test equipment, and other science-related instruments.
They have an excellent track record and reliability.
Eg BR1500VA and other models.
https://eur01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.apc.com%2F&amp;data=04%7C01%7Csylvie.le.guyader%40KI.SE%7C10fea0738af64c0d5d3c08d8d73b6687%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637495997440954136%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=1oGTEZYI047hzT5YldAD81BUKEm%2FIfEnG1FcvQpamMI%3D&amp;reserved=0
Good luck.
Colin

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Mario Methot
Sent: February 22, 2021 8:50 AM
To: [hidden email]
Subject: Anyone uses UPS units to protect your confocal/2-P systems?

*****
To join, leave or search the confocal microscopy listserv, go to:
https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&amp;data=04%7C01%7Csylvie.le.guyader%40KI.SE%7C10fea0738af64c0d5d3c08d8d73b6687%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637495997440954136%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=xwnwC4w68SIHjqvlBLSjolIQ7YU6dg6kGNhV4JUNyYk%3D&amp;reserved=0
Post images on https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&amp;data=04%7C01%7Csylvie.le.guyader%40KI.SE%7C10fea0738af64c0d5d3c08d8d73b6687%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637495997440954136%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=65UMEP7SwGjqDd9vSSAUsNWAnLTR9gae5cmqrzMrkVM%3D&amp;reserved=0 and include the link in your posting.
*****

Hi all,

We are unfortunately facing many unplanned (random) electricity outages/voltage variations in our Institute (it's a very old building) and our systems (and particularly lasers) don't like it too much... I was wondering if some of you are using UPS units to protect your confocal/2-P systems? If so what type of UPS are you using? All the info you could provide would be most welcome, thanks in advance.

Mario


När du skickar e-post till Karolinska Institutet (KI) innebär detta att KI kommer att behandla dina personuppgifter. Här finns information om hur KI behandlar personuppgifter<https://ki.se/medarbetare/integritetsskyddspolicy>.


Sending email to Karolinska Institutet (KI) will result in KI processing your personal data. You can read more about KI’s processing of personal data here<https://ki.se/en/staff/data-protection-policy>.
Douglas Richardson Douglas Richardson
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Anyone uses UPS units to protect your confocal/2-P systems?

In reply to this post by Colin Haig-2
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

We have all microscopes and processing computers on UPS for power
conditioning and emergency backup.

For 120V systems: GE VH 2KVA Tower UPS
<https://www.geupssystems.com/ge-vh-series-ups> ~$1,000.00 USD
For 209V systems: GE LP-11U UPS
<https://www.geupssystems.com/lp-11u-series-ups> ~$6,000.00 USD

-Doug

On Mon, Feb 22, 2021 at 9:08 AM Colin Haig <
[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Mario,
> We use APC Back UPS Pro sinewave units for a lot of sensitive equipment
> including CCD and CMOS cameras, test equipment, and other science-related
> instruments.
> They have an excellent track record and reliability.
> Eg BR1500VA and other models.
> https://www.apc.com
> Good luck.
> Colin
>
> -----Original Message-----
> From: Confocal Microscopy List <[hidden email]> On
> Behalf Of Mario Methot
> Sent: February 22, 2021 8:50 AM
> To: [hidden email]
> Subject: Anyone uses UPS units to protect your confocal/2-P systems?
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi all,
>
> We are unfortunately facing many unplanned (random) electricity
> outages/voltage variations in our Institute (it's a very old building) and
> our systems (and particularly lasers) don't like it too much... I was
> wondering if some of you are using UPS units to protect your confocal/2-P
> systems? If so what type of UPS are you using? All the info you could
> provide would be most welcome, thanks in advance.
>
> Mario
>
Konstantín Levitskiy Konstantín Levitskiy
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Anyone uses UPS units to protect your confocal/2-P systems?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

We have 2 units of 3000 VA (https://www.riello-ups.es/products/1-sai/44-sentinel-pro) for all components of our Leica SP2 AOBS, and only one unit for Nikon A1R+ to ptotect Argon laser and PC (in order to can save the current data).
All conventional microscope PC's have installed WinOFF programme to switch them off if 30' nobody is using them.
Safely regards,
Konstantin

________________________________
De: Confocal Microscopy List <[hidden email]> en nombre de Douglas Richardson <[hidden email]>
Enviado: lunes, 22 de febrero de 2021 15:29
Para: [hidden email] <[hidden email]>
Asunto: Re: Anyone uses UPS units to protect your confocal/2-P systems?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

We have all microscopes and processing computers on UPS for power
conditioning and emergency backup.

For 120V systems: GE VH 2KVA Tower UPS
<https://www.geupssystems.com/ge-vh-series-ups> ~$1,000.00 USD
For 209V systems: GE LP-11U UPS
<https://www.geupssystems.com/lp-11u-series-ups> ~$6,000.00 USD

-Doug

On Mon, Feb 22, 2021 at 9:08 AM Colin Haig <
[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Mario,
> We use APC Back UPS Pro sinewave units for a lot of sensitive equipment
> including CCD and CMOS cameras, test equipment, and other science-related
> instruments.
> They have an excellent track record and reliability.
> Eg BR1500VA and other models.
> https://www.apc.com
> Good luck.
> Colin
>
> -----Original Message-----
> From: Confocal Microscopy List <[hidden email]> On
> Behalf Of Mario Methot
> Sent: February 22, 2021 8:50 AM
> To: [hidden email]
> Subject: Anyone uses UPS units to protect your confocal/2-P systems?
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi all,
>
> We are unfortunately facing many unplanned (random) electricity
> outages/voltage variations in our Institute (it's a very old building) and
> our systems (and particularly lasers) don't like it too much... I was
> wondering if some of you are using UPS units to protect your confocal/2-P
> systems? If so what type of UPS are you using? All the info you could
> provide would be most welcome, thanks in advance.
>
> Mario
>
Benjamin Smith Benjamin Smith
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Anyone uses UPS units to protect your confocal/2-P systems?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

We use the following UPS setup to backup our 2-P workstations:
UPS:
https://www.cyberpowersystems.com/product/ups/smart-app-sinewave/pr2200lcdrtxl2u/

Battery extension:
https://www.cyberpowersystems.com/product/ups/extended-battery-modules/bp48v75art2u/

With this setup we can run the workstation for about 15 minutes with the
laser on (about 4 minutes without the battery extension), to give
enough time to safely save an experiment and shut everything down.  Once
the laser is on standby, the UPS can run everything else for hours.  As
others have said, the two major considerations are that you have enough VA
(2500 per high current laser system should be sufficient), and that the
output is a true sine wave, not an approximated one (approximated sine
waves can put unnecessary stress on power supplies for sensitive
equipment).

On Mon, Feb 22, 2021 at 8:13 AM MICROSCOPIA IBIS <[hidden email]>
wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> We have 2 units of 3000 VA (
> https://www.riello-ups.es/products/1-sai/44-sentinel-pro) for all
> components of our Leica SP2 AOBS, and only one unit for Nikon A1R+ to
> ptotect Argon laser and PC (in order to can save the current data).
> All conventional microscope PC's have installed WinOFF programme to switch
> them off if 30' nobody is using them.
> Safely regards,
> Konstantin
>
> ________________________________
> De: Confocal Microscopy List <[hidden email]> en nombre
> de Douglas Richardson <[hidden email]>
> Enviado: lunes, 22 de febrero de 2021 15:29
> Para: [hidden email] <[hidden email]>
> Asunto: Re: Anyone uses UPS units to protect your confocal/2-P systems?
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> We have all microscopes and processing computers on UPS for power
> conditioning and emergency backup.
>
> For 120V systems: GE VH 2KVA Tower UPS
> <https://www.geupssystems.com/ge-vh-series-ups> ~$1,000.00 USD
> For 209V systems: GE LP-11U UPS
> <https://www.geupssystems.com/lp-11u-series-ups> ~$6,000.00 USD
>
> -Doug
>
> On Mon, Feb 22, 2021 at 9:08 AM Colin Haig <
> [hidden email]> wrote:
>
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > Post images on http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > Hi Mario,
> > We use APC Back UPS Pro sinewave units for a lot of sensitive equipment
> > including CCD and CMOS cameras, test equipment, and other science-related
> > instruments.
> > They have an excellent track record and reliability.
> > Eg BR1500VA and other models.
> > https://www.apc.com
> > Good luck.
> > Colin
> >
> > -----Original Message-----
> > From: Confocal Microscopy List <[hidden email]> On
> > Behalf Of Mario Methot
> > Sent: February 22, 2021 8:50 AM
> > To: [hidden email]
> > Subject: Anyone uses UPS units to protect your confocal/2-P systems?
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > Post images on http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> > Hi all,
> >
> > We are unfortunately facing many unplanned (random) electricity
> > outages/voltage variations in our Institute (it's a very old building)
> and
> > our systems (and particularly lasers) don't like it too much... I was
> > wondering if some of you are using UPS units to protect your confocal/2-P
> > systems? If so what type of UPS are you using? All the info you could
> > provide would be most welcome, thanks in advance.
> >
> > Mario
> >
>


--
Benjamin E. Smith, Ph. D.
Imaging Specialist, Vision Science
University of California, Berkeley
195 Weill Hall
Berkeley, CA  94720-3200
Tel  (510) 642-9712
Fax (510) 643-6791
e-mail: [hidden email]
https://vision.berkeley.edu/faculty/core-grants-nei/core-grant-microscopic-imaging/
Jacqueline Ross Jacqueline Ross
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Anyone uses UPS units to protect your confocal/2-P systems?

In reply to this post by Mario Methot
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Mario,

We have a UPS for every system in our facility. They are primarily 3kVA systems, with a couple higher spec, e.g. the one for 3 phase (20kVA – over spec but was the only one available at the time).. They are mainly Socomec but we also have MGE. They are all online UPS so smooth as well as protect from spikes/outages.

Every microscope user is trained to listen for the beeping and shut the system down if this is ongoing. Also, we have them all plugged into essential power sockets. These sockets are provided with power by generators in the event of a complete campus power failure but don’t come on if it’s just local, e.g. one building/floor.

We started adding these to every new system after having had a very expensive repair bill ($10k) after our Leica confocal went off and on again. They are included in our business cases as a necessary expense.

The UPS are cheaper than the necessary repairs. We also have a contract with a local supplier to check them and replace batteries regularly,. Otherwise, they wouldn’t be much use.

They are also really good for the time-lapse systems because a short outage or spike would otherwise terminate the experiment immediately.

Kind regards,

Jacqui

From: Confocal Microscopy List <[hidden email]> On Behalf Of Mario Methot
Sent: Tuesday, February 23, 2021 2:50 AM
To: [hidden email]
Subject: Anyone uses UPS units to protect your confocal/2-P systems?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy<http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>
Post images on http://www.imgur.com<http://www.imgur.com> and include the link in your posting.
*****

Hi all,

We are unfortunately facing many unplanned (random) electricity outages/voltage variations in our Institute (it's a very old building) and our systems (and particularly lasers) don't like it too much... I was wondering if some of you are using UPS units to protect your confocal/2-P systems? If so what type of UPS are you using? All the info you could provide would be most welcome, thanks in advance.

Mario
Jacqueline Ross Jacqueline Ross
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Anyone uses UPS units to protect your confocal/2-P systems?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Mario,

One more thing I forgot to say is that our 20kVA system has an additional use as an alert system. If the power goes off, it is set up to send a signal to a phone relay system which calls our contractors and alerts them to the power outage. We have 4 phone numbers so it calls people in order and if the first 3 don’t answer, I get the call and can alert the necessary staff.

Kind regards,

Jacqui

From: Confocal Microscopy List <[hidden email]> On Behalf Of Jacqueline Ross
Sent: Tuesday, February 23, 2021 12:09 PM
To: [hidden email]
Subject: Re: Anyone uses UPS units to protect your confocal/2-P systems?

Caution - Forged Internal Domain!
This e-mail cannot be validated and may not have been sent by the sender shown in the 'From' field.
If you were not expecting to receive this e-mail we recommend you contact the sender to confirm that they sent it.
If you believe this email was legitimately sent, we suggest the sender notify the Staff Service Centre that it has been received as a forged (fake) e-mail.
Please contact the Staff Service Centre on extension 86000 if you require further assistance.

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy<http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>
Post images on http://www.imgur.com<http://www.imgur.com> and include the link in your posting.
*****

Hi Mario,

We have a UPS for every system in our facility. They are primarily 3kVA systems, with a couple higher spec, e.g. the one for 3 phase (20kVA – over spec but was the only one available at the time).. They are mainly Socomec but we also have MGE. They are all online UPS so smooth as well as protect from spikes/outages.

Every microscope user is trained to listen for the beeping and shut the system down if this is ongoing. Also, we have them all plugged into essential power sockets. These sockets are provided with power by generators in the event of a complete campus power failure but don’t come on if it’s just local, e.g. one building/floor.

We started adding these to every new system after having had a very expensive repair bill ($10k) after our Leica confocal went off and on again. They are included in our business cases as a necessary expense.

The UPS are cheaper than the necessary repairs. We also have a contract with a local supplier to check them and replace batteries regularly,. Otherwise, they wouldn’t be much use.

They are also really good for the time-lapse systems because a short outage or spike would otherwise terminate the experiment immediately.

Kind regards,

Jacqui

From: Confocal Microscopy List <[hidden email]<mailto:[hidden email]>> On Behalf Of Mario Methot
Sent: Tuesday, February 23, 2021 2:50 AM
To: [hidden email]<mailto:[hidden email]>
Subject: Anyone uses UPS units to protect your confocal/2-P systems?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy<http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy><http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy<http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>>
Post images on http://www.imgur.com<http://www.imgur.com><http://www.imgur.com<http://www.imgur.com>> and include the link in your posting.
*****

Hi all,

We are unfortunately facing many unplanned (random) electricity outages/voltage variations in our Institute (it's a very old building) and our systems (and particularly lasers) don't like it too much... I was wondering if some of you are using UPS units to protect your confocal/2-P systems? If so what type of UPS are you using? All the info you could provide would be most welcome, thanks in advance.

Mario
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Anyone uses UPS units to protect your confocal/2-P systems?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

As a related aside, most UPS systems use lead-acid batteries for storage.
These typically have a lifespan of 4-5 years in a typical UPS. Make sure to
replace them after about four years in service. Better quality UPS systems
will include an alert to indicate when the batteries are aging out.
Batteries also don't fare well in storage, so it is pointless to buy spares
ahead of time. Make sure you have some lead to account for ordering and
delivery time of the new batteries, and keep tabs on your vendor over time
to ensure you can still get the model of battery you need.

Craig

On Mon, Feb 22, 2021 at 4:27 PM Jacqueline Ross <[hidden email]>
wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Mario,
>
> One more thing I forgot to say is that our 20kVA system has an additional
> use as an alert system. If the power goes off, it is set up to send a
> signal to a phone relay system which calls our contractors and alerts them
> to the power outage. We have 4 phone numbers so it calls people in order
> and if the first 3 don’t answer, I get the call and can alert the necessary
> staff.
>
> Kind regards,
>
> Jacqui
>
> From: Confocal Microscopy List <[hidden email]> On
> Behalf Of Jacqueline Ross
> Sent: Tuesday, February 23, 2021 12:09 PM
> To: [hidden email]
> Subject: Re: Anyone uses UPS units to protect your confocal/2-P systems?
>
> Caution - Forged Internal Domain!
> This e-mail cannot be validated and may not have been sent by the sender
> shown in the 'From' field.
> If you were not expecting to receive this e-mail we recommend you contact
> the sender to confirm that they sent it.
> If you believe this email was legitimately sent, we suggest the sender
> notify the Staff Service Centre that it has been received as a forged
> (fake) e-mail.
> Please contact the Staff Service Centre on extension 86000 if you require
> further assistance.
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy<
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>
> Post images on http://www.imgur.com<http://www.imgur.com> and include the
> link in your posting.
> *****
>
> Hi Mario,
>
> We have a UPS for every system in our facility. They are primarily 3kVA
> systems, with a couple higher spec, e.g. the one for 3 phase (20kVA – over
> spec but was the only one available at the time).. They are mainly Socomec
> but we also have MGE. They are all online UPS so smooth as well as protect
> from spikes/outages.
>
> Every microscope user is trained to listen for the beeping and shut the
> system down if this is ongoing. Also, we have them all plugged into
> essential power sockets. These sockets are provided with power by
> generators in the event of a complete campus power failure but don’t come
> on if it’s just local, e.g. one building/floor.
>
> We started adding these to every new system after having had a very
> expensive repair bill ($10k) after our Leica confocal went off and on
> again. They are included in our business cases as a necessary expense.
>
> The UPS are cheaper than the necessary repairs. We also have a contract
> with a local supplier to check them and replace batteries regularly,.
> Otherwise, they wouldn’t be much use.
>
> They are also really good for the time-lapse systems because a short
> outage or spike would otherwise terminate the experiment immediately.
>
> Kind regards,
>
> Jacqui
>
> From: Confocal Microscopy List <[hidden email]<mailto:
> [hidden email]>> On Behalf Of Mario Methot
> Sent: Tuesday, February 23, 2021 2:50 AM
> To: [hidden email]<mailto:
> [hidden email]>
> Subject: Anyone uses UPS units to protect your confocal/2-P systems?
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy<
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy><
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy<
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>>
> Post images on http://www.imgur.com<http://www.imgur.com><
> http://www.imgur.com<http://www.imgur.com>> and include the link in your
> posting.
> *****
>
> Hi all,
>
> We are unfortunately facing many unplanned (random) electricity
> outages/voltage variations in our Institute (it's a very old building) and
> our systems (and particularly lasers) don't like it too much... I was
> wondering if some of you are using UPS units to protect your confocal/2-P
> systems? If so what type of UPS are you using? All the info you could
> provide would be most welcome, thanks in advance.
>
> Mario
>
Loralei Dewe-3 Loralei Dewe-3
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Anyone uses UPS units to protect your confocal/2-P systems?

In reply to this post by Mario Methot
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

When I ran a core facility I always has my equipment on UPS systems of
different types to protect the electrical equipment in an outage.

Loralei Dewe

On Mon, Feb 22, 2021, 6:00 AM Mario Methot <[hidden email]>
wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi all,
>
> We are unfortunately facing many unplanned (random) electricity
> outages/voltage variations in our Institute (it's a very old building) and
> our systems (and particularly lasers) don't like it too much... I was
> wondering if some of you are using UPS units to protect your confocal/2-P
> systems? If so what type of UPS are you using? All the info you could
> provide would be most welcome, thanks in advance.
>
> Mario
>
Zdenek Svindrych-2 Zdenek Svindrych-2
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Anyone uses UPS units to protect your confocal/2-P systems?

In reply to this post by Mario Methot
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Mario,
I prefer "online" (or double conversion) style UPS.
One 2P system I used to work with was backed up by a massive 6 kW GE
Digital Energy UPS (maybe an overkill, but there was an Ar-ion laser as
part of the system). Some smaller GE Digital Energy units (1 - 2 kW) were
also sold as Franek. Very robust, but rather old.
We also use GE VH-series units on some smaller equipment, like plate
readers and liquid chromatography systems. I now only have one microscope
backed up, it's a spinning disk confocal (no Ar-ion laser), and the UPS is
APC SURTA2200 (it's the "RT" series, that means "online"). I got it for
$200 and I'm happy with it.
Of note is that the microscope is connected to emergency power, and there
are more outages on emergency power than on regular outlets (once or twice
a year to test the generator and switches, cca 15 second outage to crank up
the diesel engine). So for any computerized equipment connected to the
emergency power a UPS is a must!
Best, zdenek

On Mon, Feb 22, 2021 at 9:00 AM Mario Methot <[hidden email]>
wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi all,
>
> We are unfortunately facing many unplanned (random) electricity
> outages/voltage variations in our Institute (it's a very old building) and
> our systems (and particularly lasers) don't like it too much... I was
> wondering if some of you are using UPS units to protect your confocal/2-P
> systems? If so what type of UPS are you using? All the info you could
> provide would be most welcome, thanks in advance.
>
> Mario
>


--
--
Zdenek Svindrych, Ph.D.
Research Scientist - Microscopy Imaging Specialist
Department of Biochemistry and Cell Biology
Geisel School of Medicine at Dartmouth
Sylvie Le Guyader Sylvie Le Guyader
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Anyone uses UPS units to protect your confocal/2-P systems?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi everyone

Nice discussion. So if I summarize what is needed:
- good UPS with batteries regularly checked and replaced when needed
- back up power from a diesel generator that takes a short while to take over (basically it is not enough to have the back up power without UPS)
- someone on call who can come before the generator has had enough

This is exactly the conclusion we came to when designing our new building.
But all this is necessary only if there are power cuts several times per year.
If the power cuts come seldom and there is a good surge protection system, there is no need for these expenses. This is of course to be balanced with the cost (and number) of the wasted experiments.

Since we have less than one power cut per year, we have skipped UPS, backup power and on call person and focussed on the surge system.

Med vänlig hälsning / Best regards

Sylvie

@@@@@@@@@@@@@@@@@@@@@@@@
Sylvie Le Guyader, PhD
Live Cell Imaging Facility Manager
Karolinska Institutet- Bionut Dpt
Blickagången 16,
Room 7362 (lab)/7840 (office)
14157 Huddinge, Sweden
mobile: +46 (0) 73 733 5008
LCI website
Follow our microscopy blog!

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Zdenek Svindrych
Sent: 23 February 2021 03:52
To: [hidden email]
Subject: Re: Anyone uses UPS units to protect your confocal/2-P systems?

*****
To join, leave or search the confocal microscopy listserv, go to:
https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&amp;data=04%7C01%7Csylvie.le.guyader%40KI.SE%7Cb246d7011b7c4719aac808d8d7a66971%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637496457022239820%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=MG41NQPGWIwA4pSzr4INZy8ITCv34%2FuP2S0VebKkSBQ%3D&amp;reserved=0
Post images on https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&amp;data=04%7C01%7Csylvie.le.guyader%40KI.SE%7Cb246d7011b7c4719aac808d8d7a66971%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637496457022239820%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=z33rkZt1nl8CkIqzUpW42Ko3GBKPYxNau%2BcU5IuxxTM%3D&amp;reserved=0 and include the link in your posting.
*****

Hi Mario,
I prefer "online" (or double conversion) style UPS.
One 2P system I used to work with was backed up by a massive 6 kW GE Digital Energy UPS (maybe an overkill, but there was an Ar-ion laser as part of the system). Some smaller GE Digital Energy units (1 - 2 kW) were also sold as Franek. Very robust, but rather old.
We also use GE VH-series units on some smaller equipment, like plate readers and liquid chromatography systems. I now only have one microscope backed up, it's a spinning disk confocal (no Ar-ion laser), and the UPS is APC SURTA2200 (it's the "RT" series, that means "online"). I got it for
$200 and I'm happy with it.
Of note is that the microscope is connected to emergency power, and there are more outages on emergency power than on regular outlets (once or twice a year to test the generator and switches, cca 15 second outage to crank up the diesel engine). So for any computerized equipment connected to the emergency power a UPS is a must!
Best, zdenek

On Mon, Feb 22, 2021 at 9:00 AM Mario Methot <[hidden email]>
wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists
> .umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&amp;data=04%7C01%7Cs
> ylvie.le.guyader%40KI.SE%7Cb246d7011b7c4719aac808d8d7a66971%7Cbff7eef1
> cf4b4f32be3da1dda043c05d%7C0%7C0%7C637496457022239820%7CUnknown%7CTWFp
> bGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn
> 0%3D%7C1000&amp;sdata=MG41NQPGWIwA4pSzr4INZy8ITCv34%2FuP2S0VebKkSBQ%3D
> &amp;reserved=0 Post images on
> https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&amp;data=04%7C01%7Csylvie.le.guyader%40KI.SE%7Cb246d7011b7c4719aac808d8d7a66971%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637496457022239820%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=z33rkZt1nl8CkIqzUpW42Ko3GBKPYxNau%2BcU5IuxxTM%3D&amp;reserved=0 and include the link in your posting.
> *****
>
> Hi all,
>
> We are unfortunately facing many unplanned (random) electricity
> outages/voltage variations in our Institute (it's a very old building)
> and our systems (and particularly lasers) don't like it too much... I
> was wondering if some of you are using UPS units to protect your
> confocal/2-P systems? If so what type of UPS are you using? All the
> info you could provide would be most welcome, thanks in advance.
>
> Mario
>


--
--
Zdenek Svindrych, Ph.D.
Research Scientist - Microscopy Imaging Specialist Department of Biochemistry and Cell Biology Geisel School of Medicine at Dartmouth


När du skickar e-post till Karolinska Institutet (KI) innebär detta att KI kommer att behandla dina personuppgifter. Här finns information om hur KI behandlar personuppgifter<https://ki.se/medarbetare/integritetsskyddspolicy>.


Sending email to Karolinska Institutet (KI) will result in KI processing your personal data. You can read more about KI’s processing of personal data here<https://ki.se/en/staff/data-protection-policy>.