To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear microscopists, We have a quite new confocal system (2 years old) and in some occasions we have doubts about whether our Ar ion multiline laser is experiencing some problems. The Official Technical Service tell us that it is impossible to get a new gas laser since they are no longer build. So, instead of multiline Ar laser they claim we should change to a solid one-line 488nm laser. Fortunately, in fact the problem was different (the analyzer for Nomarski contrast was put in the optic path). In summary, the official tech ensures that we have to use multiline laser switched to maximum potency. So, even though we do not use much any other lines, different than 488nm, nevertheless we would like to prevent the early aging of our Ar laser, which would likely happen if we stay using it at maximum potency. Is really the use at maximum power the best way to prolong the life time for Ar laser? Or it would be preferable work at, for example 50%? Best regards, Dr. Konstantín Levitskiy Servicio de Microscopía InstitutodeBiomedicinadeSevilla - IBiS Campus del Hospital Universitario Virgen del Rocío Avda. Manuel Siurot s/nº 41013 Sevilla Tlfno: 955 92 3030 Email: <mailto:[hidden email]> [hidden email] Web: <http://www.ibis-sevilla.es/> www.ibis-sevilla.es |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Konstantin, I am going through exactly the same problem right now. I have a confocal system with a dead argon-ion laser and I am working on replacements. In the past, I have had my laser 'refurbished'. I don't know what your repair options are in Spain, but I was able to send my laser to a company in the US to have it re-tubed. I have done this two or three times now at a cost of about $5000 USD for each refurbishment. The laser comes back 'like new' after this as the company replaces the expired gas tube and checks all the electronics and optics. The only problem is that Argon lasers seeing heavy use are typically only going to last a few years at best, so the cost of sending it out for repair (not to mention system down time while the laser is being shipped to the USA) is starting to be a problem. We have started to move towards work that only requires the 488nm line of the argon (previously the 457nm line was important to us, but not anymore) so I am considering replacing our Argon with a direct diode laser. This should have about 10x to 20x the lifespan of a typical Argon laser so will save time and money in the long run. Also, for 488nm lasers of modest specifications the pricing isn't much more than what I was paying to have the Argon laser refurbished. I am currently shopping around, so if anyone is interested in my results please email me privately in a couple weeks. Craig On Thu, Nov 2, 2017 at 9:11 AM, Konstantín Levitskiy <[hidden email] > wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear microscopists, > > We have a quite new confocal system (2 years old) and in some occasions we > have doubts about whether our Ar ion multiline laser is experiencing some > problems. The Official Technical Service tell us that it is impossible to > get a new gas laser since they are no longer build. So, instead of > multiline > Ar laser they claim we should change to a solid one-line 488nm laser. > Fortunately, in fact the problem was different (the analyzer for Nomarski > contrast was put in the optic path). > > In summary, the official tech ensures that we have to use multiline laser > switched to maximum potency. So, even though we do not use much any other > lines, different than 488nm, nevertheless we would like to prevent the > early > aging of our Ar laser, which would likely happen if we stay using it at > maximum potency. Is really the use at maximum power the best way to prolong > the life time for Ar laser? Or it would be preferable work at, for example > 50%? > > Best regards, > > Dr. Konstantín Levitskiy > > Servicio de Microscopía > > InstitutodeBiomedicinadeSevilla - IBiS > > Campus del Hospital Universitario Virgen del Rocío > > Avda. Manuel Siurot s/nº > > 41013 Sevilla > > Tlfno: 955 92 3030 > > Email: <mailto:[hidden email]> [hidden email] > > Web: <http://www.ibis-sevilla.es/> www.ibis-sevilla.es > > > |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Konstantin, My experience with several argon lasers is that when you need to increase current to max the laser tube is approaching its end. The higher current you set, the more stable will be the emission, however this shortens lifetime (e.g. in standby the current is set at ~3A or ~40%).To prolong lifetime run laser at 50% if the emitted power and stability is aceptable. I am surprised to hear somebody stating that Ar lasers are not available anymore. At least some companies like Lasos are still making them (no commercial interest). I replaced Lasos Ar multiline laser for below $5000 two years ago (just laser head keeoing the old power supply). For the same price it is better than the refurbishment proposed by Craig. Best, Arvydas Arvydas Matiukas, Ph.D. Director of Confocal&Two-Photon Core SUNY Upstate Medical University Syracuse, USA >>> Craig Brideau <[hidden email]> 11/2/2017 1:54 PM >>> ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Konstantin, I am going through exactly the same problem right now. I have a confocal system with a dead argon-ion laser and I am working on replacements. In the past, I have had my laser 'refurbished'. I don't know what your repair options are in Spain, but I was able to send my laser to a company in the US to have it re-tubed. I have done this two or three times now at a cost of about $5000 USD for each refurbishment. The laser comes back 'like new' after this as the company replaces the expired gas tube and checks all the electronics and optics. The only problem is that Argon lasers seeing heavy use are typically only going to last a few years at best, so the cost of sending it out for repair (not to mention system down time while the laser is being shipped to the USA) is starting to be a problem. We have started to move towards work that only requires the 488nm line of the argon (previously the 457nm line was important to us, but not anymore) so I am considering replacing our Argon with a direct diode laser. This should have about 10x to 20x the lifespan of a typical Argon laser so will save time and money in the long run. Also, for 488nm lasers of modest specifications the pricing isn't much more than what I was paying to have the Argon laser refurbished. I am currently shopping around, so if anyone is interested in my results please email me privately in a couple weeks. Craig On Thu, Nov 2, 2017 at 9:11 AM, Konstantín Levitskiy <[hidden email] > wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear microscopists, > > We have a quite new confocal system (2 years old) and in some occasions we > have doubts about whether our Ar ion multiline laser is experiencing some > problems. The Official Technical Service tell us that it is impossible to > get a new gas laser since they are no longer build. So, instead of > multiline > Ar laser they claim we should change to a solid one-line 488nm laser. > Fortunately, in fact the problem was different (the analyzer for Nomarski > contrast was put in the optic path). > > In summary, the official tech ensures that we have to use multiline laser > switched to maximum potency. So, even though we do not use much any other > lines, different than 488nm, nevertheless we would like to prevent the > early > aging of our Ar laser, which would likely happen if we stay using it at > maximum potency. Is really the use at maximum power the best way to prolong > the life time for Ar laser? Or it would be preferable work at, for example > 50%? > > Best regards, > > Dr. Konstantín Levitskiy > > Servicio de Microscopía > > InstitutodeBiomedicinadeSevilla - IBiS > > Campus del Hospital Universitario Virgen del Rocío > > Avda. Manuel Siurot s/n* > > 41013 Sevilla > > Tlfno: 955 92 3030 > > Email: <mailto:[hidden email]> [hidden email] > > Web: <http://www.ibis-sevilla.es/> www.ibis-sevilla.es > > > |
![]() ![]() |
Nikos Ekizoglou - Planelight |
![]() |
In reply to this post by Konstantín Levitskiy
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Konstantin, The Ar lasers have not disappeared but they are replaced more and more with multilaser light engines that are using diode lasers. You can still buy a new Ar laser but for sure the first step is to refurbish the one you already have. What they normally do is recharge the gas and in the best case they change the whole gas tube. This normally comes with one year warranty and it is much cheaper than buying a new Ar laser. If you are in Spain and you are interested in refurbishing your laser, send me an email and I can send you the contact details. One of the possible reasons that the Ar lasers are not lasting long is that they are used at full power all the time while the microscope controls the laser line and intensity of light reaching the sample by AOTF modules. Best regards Nikos Ekizoglou Plane Light S.L. C/ Riocabado, 4 28047 Madrid (Spain) Phone: +34 911 130 824 Cell: +34 650 70 52 39 Fax: +34 910 113 757 www.planelight.net -----Mensaje original----- De: Confocal Microscopy List [mailto:[hidden email]] En nombre de Konstantín Levitskiy Enviado el: jueves, 02 de noviembre de 2017 16:11 Para: [hidden email] Asunto: Ar ion laser durability ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear microscopists, We have a quite new confocal system (2 years old) and in some occasions we have doubts about whether our Ar ion multiline laser is experiencing some problems. The Official Technical Service tell us that it is impossible to get a new gas laser since they are no longer build. So, instead of multiline Ar laser they claim we should change to a solid one-line 488nm laser. Fortunately, in fact the problem was different (the analyzer for Nomarski contrast was put in the optic path). In summary, the official tech ensures that we have to use multiline laser switched to maximum potency. So, even though we do not use much any other lines, different than 488nm, nevertheless we would like to prevent the early aging of our Ar laser, which would likely happen if we stay using it at maximum potency. Is really the use at maximum power the best way to prolong the life time for Ar laser? Or it would be preferable work at, for example 50%? Best regards, Dr. Konstantín Levitskiy Servicio de Microscopía InstitutodeBiomedicinadeSevilla - IBiS Campus del Hospital Universitario Virgen del Rocío Avda. Manuel Siurot s/nº 41013 Sevilla Tlfno: 955 92 3030 Email: <mailto:[hidden email]> [hidden email] Web: <http://www.ibis-sevilla.es/> www.ibis-sevilla.es |
![]() |
In reply to this post by Konstantín Levitskiy
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Konstantin, you might get more specific answers if you would mention which system you are talking about. And you might be interested to hear that I got an offer for replacement of an Argon laser as of last spring. I believe they still should be available for repairs. Anyway, on our old Leica SP5 we operated the Argon laser always at 20 % hardware power. We were told that below 15 % the laser output would fluctuate quite a bit so 20 % seemed like a safe margin. Despite this setting, we didn't need to set the AOTF to more than 30 % (usually <10%). We did not use it at full power specifically to prolong life time. So the only experiment I could imagine where it would make sense to crank up the hardware power on that old system would be stuff like FRAP. You might have a different Argon laser though and stability might be different (and apart from stability I do not see what could be the problem with a laser operated at reduced power. Anyone?). If you worry about stability you could do a test scan with a fluorescent plastic slide, like the ones from Chroma or some other uniformly fluorescent test sample. If you then scan with a low res objective and use several detectors in parallel (including transmission if available) then you can be pretty sure that any change in intensity seen in all detectors is due to laser intensity issues (and not to detector problems). If you compare the result for full hardware power and low AOTF setting to lower hardware power and more open AOTF (line to line as well as long term and ideally day to day) you should get the answer if usage of your laser at lower power is ok. (Make sure the laser had enough time to equilibrate at each power setting. Fluctuation is supposed to be worse during warm up.) Good luck Steffen Am 02.11.2017 um 16:11 schrieb Konstantín Levitskiy: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear microscopists, > > We have a quite new confocal system (2 years old) and in some occasions we > have doubts about whether our Ar ion multiline laser is experiencing some > problems. The Official Technical Service tell us that it is impossible to > get a new gas laser since they are no longer build. So, instead of multiline > Ar laser they claim we should change to a solid one-line 488nm laser. > Fortunately, in fact the problem was different (the analyzer for Nomarski > contrast was put in the optic path). > > In summary, the official tech ensures that we have to use multiline laser > switched to maximum potency. So, even though we do not use much any other > lines, different than 488nm, nevertheless we would like to prevent the early > aging of our Ar laser, which would likely happen if we stay using it at > maximum potency. Is really the use at maximum power the best way to prolong > the life time for Ar laser? Or it would be preferable work at, for example > 50%? > > Best regards, > > Dr. Konstantín Levitskiy > > Servicio de Microscopía > > InstitutodeBiomedicinadeSevilla - IBiS > > Campus del Hospital Universitario Virgen del Rocío > > Avda. Manuel Siurot s/nº > > 41013 Sevilla > > Tlfno: 955 92 3030 > > Email: <mailto:[hidden email]> [hidden email] > > Web: <http://www.ibis-sevilla.es/> www.ibis-sevilla.es > > > -- -- ---------------------------------------------------------- Steffen Dietzel, PD Dr. rer. nat. Head of the Core Facility Bioimaging at the Biomedical Center Ludwig-Maximilians-Universität München Walter-Brendel-Zentrum für Experimentelle Medizin Address: Biomedical Center Großhaderner Straße 9 D-82152 Planegg-Martinsried Phone: +49/89/2180-71518 skype: steffendietzel e-mail: [hidden email] fax-to-e-mail: +49/89/2180-9971518 http://www.bioimaging.bmc.med.uni-muenchen.de -- ------------------------------------------------------------ Steffen Dietzel, PD Dr. rer. nat Ludwig-Maximilians-Universität München Biomedical Center (BMC) Head of the Core Facility Bioimaging Großhaderner Straße 9 D-82152 Planegg-Martinsried Germany http://www.bioimaging.bmc.med.uni-muenchen.de |
![]() ![]() |
Konstantín Levitskiy |
![]() |
In reply to this post by Konstantín Levitskiy
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi everybody, thanks all of you for your advices. As I mentioned, our laser (Melles Griot) issue has been resolved, there was no trouble with it. So my intention is to find out the way to prolong the laser useful lifetime. And yes, we do have a knob for the laser power supply. We'll follow your advices to get the minimum power current with a maximum stability. "TO SERVE AND PROTECT". Best regards, DR. KONSTANTÍN LEVITSKY SERVICIO DE MICROSCOPÍA INSTITUTODEBIOMEDICINADESEVILLA - IBIS CAMPUS DEL HOSPITAL UNIVERSITARIO VIRGEN DEL ROCÍO AVDA. MANUEL SIUROT S/Nº 41013 SEVILLA TLFNO: 955 92 3030 EMAIL: [hidden email] WEB: WWW.IBIS-SEVILLA.ES [1] Links: ------ [1] http://www.ibis-sevilla.es |
![]() |
In reply to this post by Konstantín Levitskiy
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** A number of 'listers have contacted me asking for details of my 488 laser replacement saga. I've received quotes from a few vendors and all were quite reasonable. I won't name names in the interest of fairness, but in general, the quotes came in between about $5500 and $8000 USD. The price difference mainly scaled with the power of the laser, with ~20-25mW on the cheap end and >100mW on the high end. Please note this was for a North American purchase (Canada) and I specified single spatial-mode TEM00 with wavelength stabilization not required. I was pretty impressed with all the quotes I received as the feature sets offered by all the vendors were pretty similar (includes all three of analog control, TTL control, manual control). Cost basically scales over the stated range with maximum power output. Expect variations for your locale and be sure to contact your local representatives for pricing in your area. Import fees and handling charges can be significant so this will lead to pricing differences. Craig On Thu, Nov 2, 2017 at 1:21 PM, Nikos Ekizoglou - Planelight < [hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear Konstantin, > The Ar lasers have not disappeared but they are replaced more and more with > multilaser light engines that are using diode lasers. You can still buy a > new Ar laser but for sure the first step is to refurbish the one you > already > have. What they normally do is recharge the gas and in the best case they > change the whole gas tube. This normally comes with one year warranty and > it > is much cheaper than buying a new Ar laser. If you are in Spain and you are > interested in refurbishing your laser, send me an email and I can send you > the contact details. > One of the possible reasons that the Ar lasers are not lasting long is that > they are used at full power all the time while the microscope controls the > laser line and intensity of light reaching the sample by AOTF modules. > Best regards > > Nikos Ekizoglou > > Plane Light S.L. > C/ Riocabado, 4 > 28047 Madrid (Spain) > Phone: +34 911 130 824 > Cell: +34 650 70 52 39 > Fax: +34 910 113 757 > www.planelight.net > > > > -----Mensaje original----- > De: Confocal Microscopy List [mailto:[hidden email]] En > nombre de Konstantín Levitskiy > Enviado el: jueves, 02 de noviembre de 2017 16:11 > Para: [hidden email] > Asunto: Ar ion laser durability > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear microscopists, > > We have a quite new confocal system (2 years old) and in some occasions we > have doubts about whether our Ar ion multiline laser is experiencing some > problems. The Official Technical Service tell us that it is impossible to > get a new gas laser since they are no longer build. So, instead of > multiline > Ar laser they claim we should change to a solid one-line 488nm laser. > Fortunately, in fact the problem was different (the analyzer for Nomarski > contrast was put in the optic path). > > In summary, the official tech ensures that we have to use multiline laser > switched to maximum potency. So, even though we do not use much any other > lines, different than 488nm, nevertheless we would like to prevent the > early > aging of our Ar laser, which would likely happen if we stay using it at > maximum potency. Is really the use at maximum power the best way to prolong > the life time for Ar laser? Or it would be preferable work at, for example > 50%? > > Best regards, > > Dr. Konstantín Levitskiy > > Servicio de Microscopía > > InstitutodeBiomedicinadeSevilla - IBiS > > Campus del Hospital Universitario Virgen del Rocío > > Avda. Manuel Siurot s/nº > > 41013 Sevilla > > Tlfno: 955 92 3030 > > Email: <mailto:[hidden email]> [hidden email] > > Web: <http://www.ibis-sevilla.es/> www.ibis-sevilla.es > > > |