Dear all,
A labmate of mine is contemplating bandpass filter-set (25nm FWHM with center wavelengths of 450 to 675nm at intervals of 25 nm) for doing spectral unmixing. To go with the filters he is looking for 10 position motorized wheel. Preferably the filters (and wheel adapters) should have 1" diameter, but smaller diameter is also usable. CVI does provide bandpass filters with 25nm FWHM but only at 50nm intervals (whereas we need 25nm interval). We are yet to hear about custom filters for in-between center wavelengths. Are there any other off-the-shelf high transmission filter sets available from other vendors with said bandwidths? For filter wheel, we got a quote from Ludl for 10 positions wheel of around 5.5K USD. Thorlabs is selling 6 position filter wheel at approx. 1K USD. But unfortunately they do not have 10 position wheels. Ludl filter wheels are switchable in 50ns whereas thorlabs' in 300ms. But he doesn't need real high speeds. Any other vendors of 10 position wheels that cost say, 1.5K? Thanks for your suggestions. Commercial responses are welcome off the list. Regards, Shalin -- My co-ordinates: Shalin Mehta, Graduate student Graduate Programme in Bioengineering, NUS, Singapore Mobile: +65 90694182 |
![]() ![]() |
Michael Weber-4 |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear Shalin, what about two filter wheels with six positions in a row? Leave one position empty, and then you have your 10 positions. The bandpass filters can be custom-made by Chroma, but that's probably not cheap. cheers, Michael Shalin Mehta wrote: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear all, > > A labmate of mine is contemplating bandpass filter-set (25nm FWHM with > center wavelengths of 450 to 675nm at intervals of 25 nm) for doing > spectral unmixing. > > To go with the filters he is looking for 10 position motorized wheel. > Preferably the filters (and wheel adapters) should have 1" diameter, but > smaller diameter is also usable. CVI does provide bandpass filters with > 25nm FWHM but only at 50nm intervals (whereas we need 25nm interval). We > are yet to hear about custom filters for in-between center wavelengths. > > Are there any other off-the-shelf high transmission filter sets > available from other vendors with said bandwidths? > > For filter wheel, we got a quote from Ludl for 10 positions wheel of > around 5.5K USD. Thorlabs is selling 6 position filter wheel at approx. > 1K USD. But unfortunately they do not have 10 position wheels. Ludl > filter wheels are switchable in 50ns whereas thorlabs' in 300ms. But he > doesn't need real high speeds. Any other vendors of 10 position wheels > that cost say, 1.5K? > > Thanks for your suggestions. > Commercial responses are welcome off the list. > > Regards, > Shalin > > > > -- > My co-ordinates: > Shalin Mehta, Graduate student > Graduate Programme in Bioengineering, NUS, Singapore > Mobile: +65 90694182 ... [show rest of quote]
|
![]() ![]() |
George McNamara |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Shalin Mehta
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Shalin, Sutter's Lambda-10 series fits your needs, though not your price (price list is at www.sutter.com). They have controllers for 1, 2 or 3 wheels, and 1, or more, shutters. The Sutter products can be controlled by most software packages. Prior Instruments and Applied Scientific Instruments sell wheels, but probably not at the price point you want. You already mentioned Ludl. With respect to ThorLabs, two 6-position wheels, in series, each with one empty position = 10 filter positions, but at the risk of vignetting. Chroma, Omega and Semrock have plenty of filters. You can look at SBIG (www.sbig.com) to see if they have any 10 position wheels, may hit your price point but you may have a lot of work to get their macro language to work with your system. If this is on a motorized Zeiss stand, ask your local rep about Zeiss' "Pinkel" wheel. Price point may be close to what you want. Yuval Garini and Ken Castleman had articles in Cytometry discussing (ok, arguing) the merits of narrow vs wide filter based approaches. Having worked with Yuval, I lean toward his suggestion of 'wider is better'. This would also cover a wider spectral range, i.e. from the new blue to far-red fluorescent proteins. You may also want to take a look at Michael Speicher's 8-color M-FISH paper(s), since that is probably the record for the number of fluorophores in a light microscope experiment. Both SKY (Yuval) and M-FISH (Castleman and Speicher) have been unmixing 24 or more combinatorial targets for a decade. At 02:41 AM 11/14/2007, you wrote: >Search the CONFOCAL archive at >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear all, > > A labmate of mine is contemplating bandpass filter-set (25nm FWHM > with center wavelengths of 450 to 675nm at intervals of 25 nm) for > doing spectral unmixing. > >To go with the filters he is looking for 10 position motorized >wheel. Preferably the filters (and wheel adapters) should have 1" >diameter, but smaller diameter is also usable. CVI does provide >bandpass filters with 25nm FWHM but only at 50nm intervals (whereas >we need 25nm interval). We are yet to hear about custom filters for >in-between center wavelengths. > >Are there any other off-the-shelf high transmission filter sets >available from other vendors with said bandwidths? > >For filter wheel, we got a quote from Ludl for 10 positions wheel of >around 5.5K USD. Thorlabs is selling 6 position filter wheel at >approx. 1K USD. But unfortunately they do not have 10 position >wheels. Ludl filter wheels are switchable in 50ns whereas thorlabs' >in 300ms. But he doesn't need real high speeds. Any other vendors of >10 position wheels that cost say, 1.5K? > >Thanks for your suggestions. >Commercial responses are welcome off the list. > >Regards, >Shalin > > > >-- >My co-ordinates: >Shalin Mehta, Graduate student >Graduate Programme in Bioengineering, NUS, Singapore >Mobile: +65 90694182 ... [show rest of quote] George McNamara, Ph.D. University of Miami, Miller School of Medicine Image Core Miami, FL 33010 [hidden email] [hidden email] 305-243-8436 office http://home.earthlink.net/~pubspectra/ http://home.earthlink.net/~geomcnamara/ http://www.sylvester.org/health_pro/shared_resources/index.asp (see Analytical Imaging Core Facility) |
![]() ![]() |
Shalin Mehta |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Thanks a lot for pointers George and Michael.
Regards, Shalin On Nov 14, 2007 7:53 PM, George McNamara <[hidden email]> wrote:
Hi Shalin, ... [show rest of quote] -- My co-ordinates: Shalin Mehta, Graduate student Graduate Programme in Bioengineering, NUS, Singapore Mobile: +65 90694182 |
![]() ![]() |
J.P. Shields |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by George McNamara
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi, I am having problems with fluctuating voltage since our physical plant replaced the perfectly good transformer outside the building with a "new" version (my wife works at a transformer plant, so I know they don't go bad unless something catastrophic occurs). Since then I've had some pretty good spikes, surges, etc.. and most are correlated with increase/decrease of student presence on campus. Our campus electrical experts and confocal service engineer are better at pointing fingers than providing solutions. I feel better now. On to the important stuff. I was hoping for some feedback on good line conditioners that would be appropriate for a variety of imaging equipment. I realize that several different configurations might be needed. We have a confocal and two SEMs affected. I suspect our eventual, new confocal will experience similar difficulties. Thanks in advance! John Shields EM Lab University of Georgia Athens, GA 30602 [hidden email] |
![]() ![]() |
Shaw, Andrew-2 |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal We've been very happy with our UPS / Line conditioner from EATON Powerware. Our building has a history of power fluctuations, etc. The unit is a little noisier (louder, not noisy power) than I'd like, but it's housed in the same room with our Leica SP5 and still the room is not too noisy too work in. They have multiple sizes depending upon the power requirements of the systems you'll be driving. Andrew Shaw Lab Coordinator Cell and Tissue Imaging Core Morehouse School of Medicine Atlanta, GA 30310 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of J.P. Shields Sent: Tuesday, November 27, 2007 4:09 PM To: [hidden email] Subject: Line conditioners Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi, I am having problems with fluctuating voltage since our physical plant replaced the perfectly good transformer outside the building with a "new" version (my wife works at a transformer plant, so I know they don't go bad unless something catastrophic occurs). Since then I've had some pretty good spikes, surges, etc.. and most are correlated with increase/decrease of student presence on campus. Our campus electrical experts and confocal service engineer are better at pointing fingers than providing solutions. I feel better now. On to the important stuff. I was hoping for some feedback on good line conditioners that would be appropriate for a variety of imaging equipment. I realize that several different configurations might be needed. We have a confocal and two SEMs affected. I suspect our eventual, new confocal will experience similar difficulties. Thanks in advance! John Shields EM Lab University of Georgia Athens, GA 30602 [hidden email] |
![]() ![]() |
Boswell, Carl A - (cboswell) |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by J.P. Shields
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi John, I also have had the opportunity to use a couple of Powerware units, and they have been very dependable. The first unit, installed on a Nikon PCM2000 confocal solved many mysterious software gliches. Zeiss included another brand UPS unit in their quote for our 510 Meta that was very expensive. They claimed it was the only unit that could do the job. As an alternative, I went to Powerware where the folks there were very helpful, understood what was required and provided exactly the configuration I needed. There have been no problems (with the power at least) since it was installed almost two years ago. Cheers, Carl Carl A. Boswell, Ph.D. Molecular and Cellular Biology University of Arizona 520-954-7053 FAX 520-621-3709 ----- Original Message ----- From: "J.P. Shields" <[hidden email]> To: <[hidden email]> Sent: Tuesday, November 27, 2007 2:09 PM Subject: Line conditioners > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Hi, > I am having problems with fluctuating voltage since our physical plant > replaced the perfectly good transformer outside the building with a "new" > version (my wife works at a transformer plant, so I know they don't go bad > unless something catastrophic occurs). Since then I've had some pretty > good > spikes, surges, etc.. and most are correlated with increase/decrease of > student presence on campus. Our campus electrical experts and confocal > service engineer are better at pointing fingers than providing solutions. > I feel better now. On to the important stuff. > > I was hoping for some feedback on good line conditioners that would be > appropriate for a variety of imaging equipment. I realize that several > different configurations might be needed. We have a confocal and two SEMs > affected. I suspect our eventual, new confocal will experience similar > difficulties. > > Thanks in advance! > John Shields > EM Lab > University of Georgia > Athens, GA 30602 > [hidden email] > ... [show rest of quote]
|
![]() ![]() |
David Burk |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by J.P. Shields
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal John, We got two of these critters when I was still at the LSU main campus: http://www.tripplite.com/products/product.cfm?productID=1972 to fulfill Leica's requirements to protect our SP2. We had horrible line conditions that caused some hardware failures and these fixed the problems. Kind of pricy though. David [hidden email] -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of J.P. Shields Sent: Tuesday, November 27, 2007 3:09 PM To: [hidden email] Subject: Line conditioners Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi, I am having problems with fluctuating voltage since our physical plant replaced the perfectly good transformer outside the building with a "new" version (my wife works at a transformer plant, so I know they don't go bad unless something catastrophic occurs). Since then I've had some pretty good spikes, surges, etc.. and most are correlated with increase/decrease of student presence on campus. Our campus electrical experts and confocal service engineer are better at pointing fingers than providing solutions. I feel better now. On to the important stuff. I was hoping for some feedback on good line conditioners that would be appropriate for a variety of imaging equipment. I realize that several different configurations might be needed. We have a confocal and two SEMs affected. I suspect our eventual, new confocal will experience similar difficulties. Thanks in advance! John Shields EM Lab University of Georgia Athens, GA 30602 [hidden email] |