Ca-measurement in glandular tissue

classic Classic list List threaded Threaded
5 messages Options
Cselenyak Attila Cselenyak Attila
Reply | Threaded
Open this post in threaded view
|

Ca-measurement in glandular tissue

Dear List members,

we would like measure Ca(2+) in glandular tissue with Fluo3 using
confocal mircoscopy. We tried to load the tissue pieces with different
concentrations of the dye for 1,5 hours in 37C, washing after that
with PBS, and the tissue was put in a perfusion system. But we could
not see any changes in the fluorescence intensity when giving either
ATP or A23187 ionophore to the perfusion system.

Do you have any experience what would be an optimal loading protocol
or Fluo3 is suitable for this kind of measurements?

Thanks is advence!

Best regards,

Attila

--
Attila Cselenyak
PhD student

Semmelweis University
Basic Medical Science Center
Institute Of Human Physiology And Clinical Experimental Research
H-1094, Budapest
Tűzoltó utca 37-47.
Phone: +36-1-4591500 ext.60342
Mobile: +36-70-3874931
Axel Kurt Preuss Axel Kurt Preuss
Reply | Threaded
Open this post in threaded view
|

Re: Ca-measurement in glandular tissue

Dear Attila
1.5 hrs is far too long
I would try 40 min, maybe even only 30 min
Then washing in HBSS. The phosphate in PBS can be a problem (catching
your Ca away)
I would clamp Ca down to 1 mM , maybe even 300 micro M during washing

But the most critical step is what cells you want to label. If you
soak the whole tissue in fluo 4 you might get abundant fluorescence
making cells indiscriminable or buffer too much Ca



Cheers

Axel

Axel K Preuss PhD, A*Star IMCB-Central Imaging Facility 6-19B   Sent
from. +65 9271 5622

On May 11, 2010, at 5:50 PM, "Cselenyak Attila" <[hidden email]>
wrote:

> Dear List members,
>
> we would like measure Ca(2+) in glandular tissue with Fluo3 using
> confocal mircoscopy. We tried to load the tissue pieces with different
> concentrations of the dye for 1,5 hours in 37C, washing after that
> with PBS, and the tissue was put in a perfusion system. But we could
> not see any changes in the fluorescence intensity when giving either
> ATP or A23187 ionophore to the perfusion system.
>
> Do you have any experience what would be an optimal loading protocol
> or Fluo3 is suitable for this kind of measurements?
>
> Thanks is advence!
>
> Best regards,
>
> Attila
>
> --
> Attila Cselenyak
> PhD student
>
> Semmelweis University
> Basic Medical Science Center
> Institute Of Human Physiology And Clinical Experimental Research
> H-1094, Budapest
> Tűzoltó utca 37-47.
> Phone: +36-1-4591500 ext.60342
> Mobile: +36-70-3874931

Note: This message may contain confidential information. If this Email/Fax has been sent to you by mistake, please notify the sender and delete it immediately. Thank you.
Louis Villeneuve Louis Villeneuve
Reply | Threaded
Open this post in threaded view
|

RE Ca-measurement in glandular tissue

In reply to this post by Cselenyak Attila

Hi Attila,

Do you see any fluorescence? Does fluo3 etting inside the tissue?  Maybe adding some pluronic acid can help to have a better signal!

Have a nice day,

Louisssssssss


Cselenyak Attila <[hidden email]>@LISTS.UMN.EDU
Envoyé par : Confocal Microscopy List <[hidden email]>

2010-05-11 05:41

Veuillez répondre à
Confocal Microscopy List <[hidden email]>

A
[hidden email]
cc
Objet
Ca-measurement in glandular tissue







Dear List members,

we would like measure Ca(2+) in glandular tissue with Fluo3 using
confocal mircoscopy. We tried to load the tissue pieces with different
concentrations of the dye for 1,5 hours in 37C, washing after that
with PBS, and the tissue was put in a perfusion system. But we could
not see any changes in the fluorescence intensity when giving either
ATP or A23187 ionophore to the perfusion system.

Do you have any experience what would be an optimal loading protocol
or Fluo3 is suitable for this kind of measurements?

Thanks is advence!

Best regards,

Attila

--
Attila Cselenyak
PhD student

Semmelweis University
Basic Medical Science Center
Institute Of Human Physiology And Clinical Experimental Research
H-1094, Budapest
Tűzoltó utca 37-47.
Phone: +36-1-4591500 ext.60342
Mobile: +36-70-3874931



Axel Kurt Preuss Axel Kurt Preuss
Reply | Threaded
Open this post in threaded view
|

Re: RE Ca-measurement in glandular tissue

He is compartmentalising his fluo, so whether he sees fluorescence or notay be secondary
After 1.5 hrs loading time the poor fluo 4 molecules are probably  gasping in pain and can't take it anymore ( with it I mean the Ca) (scientifically actually they probably could "take" the Ca but without signal increase

Best to load cells the old fashioned  way per injection.

But If it has to be AM  best to try different concentrations
 in incubator for not more than 40 minutes
Yes, with so many cells in the tissue competing for fluo it could be fluo concentration is not sufficient. But he HAS to try different concentrations, as overloading could also be the case, and MUST avoid compartmentalisation

And I would stay away from PBS

It is aost  hopeless to try to "stain" tissue fragments with Ca dye (AM)
Bbbwouldb probably only work if there are cells extruding and not touching other tissue. Such exposed  cells might take up fluo the right way

Best to use a method to selectively target a few cells. This means injecting the free acid or experimenting with electrophoresis (dextran coupled dyes) or local perfusion with AM

Or to digest the tissue to the point of discrete cells

My text may be garbled due to compartmentilizing iPhone keyboard









Thanks

Axel

Axel K Preuss PhD, A*Star IMCB-Central Imaging Facility 6-19B   Sent from. +65 9271 5622

On May 11, 2010, at 8:38 PM, "[hidden email]<mailto:[hidden email]>" <[hidden email]<mailto:[hidden email]>> wrote:


Hi Attila,

Do you see any fluorescence? Does fluo3 etting inside the tissue?  Maybe adding some pluronic acid can help to have a better signal!

Have a nice day,

Louisssssssss


Cselenyak Attila <[hidden email]<mailto:[hidden email]>>@LISTS.UMN.EDU
Envoyé par : Confocal Microscopy List <[hidden email]<mailto:[hidden email]>>

2010-05-11 05:41

Veuillez répondre à
Confocal Microscopy List <[hidden email]<mailto:[hidden email]>>



A
        [hidden email]<mailto:[hidden email]>
cc

Objet
        Ca-measurement in glandular tissue









Dear List members,

we would like measure Ca(2+) in glandular tissue with Fluo3 using
confocal mircoscopy. We tried to load the tissue pieces with different
concentrations of the dye for 1,5 hours in 37C, washing after that
with PBS, and the tissue was put in a perfusion system. But we could
not see any changes in the fluorescence intensity when giving either
ATP or A23187 ionophore to the perfusion system.

Do you have any experience what would be an optimal loading protocol
or Fluo3 is suitable for this kind of measurements?

Thanks is advence!

Best regards,

Attila

--
Attila Cselenyak
PhD student

Semmelweis University
Basic Medical Science Center
Institute Of Human Physiology And Clinical Experimental Research
H-1094, Budapest
Tűzoltó utca 37-47.
Phone: +36-1-4591500 ext.60342
Mobile: +36-70-3874931




Note: This message may contain confidential information. If this Email/Fax has been sent to you by mistake, please notify the sender and delete it immediately. Thank you.
Periasamy, Ammasi (ap3t) Periasamy, Ammasi (ap3t)
Reply | Threaded
Open this post in threaded view
|

Re: RE Ca-measurement in glandular tissue

In reply to this post by Louis Villeneuve

We did collect calcium signal in the tissue using calcium green, but we injected the dye. We are getting pretty good signals.

Hope this helps.

 

Ammasi Periasamy, Ph.D.

Director, Keck Center for Cellular Imaging (KCCI)

Professor of Biology and Biomedical Engineering

Biology, Gilmer Hall (064), 485 McCormick Rd

University of Virginia

Charlottesville, VA 22904

Voice: 434-243-7602 (Office); 982-4869 (lab)

Fax:434-982-5210; Email:[hidden email]

http://www.kcci.virginia.edu

************************

10th Annual Workshop on FRET Microscopy, March 8-12, 2011

http://www.kcci.virginia.edu/workshop/workshop2011/index.php

*************************

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email]
Sent: Tuesday, May 11, 2010 8:42 AM
To: [hidden email]
Subject: RE Ca-measurement in glandular tissue

 


Hi Attila,

Do you see any fluorescence? Does fluo3 etting inside the tissue?  Maybe adding some pluronic acid can help to have a better signal!

Have a nice day,

Louisssssssss

Cselenyak Attila <[hidden email]>@LISTS.UMN.EDU
Envoyé par : Confocal Microscopy List <[hidden email]>

2010-05-11 05:41

Veuillez répondre à
Confocal Microscopy List <[hidden email]>

A

[hidden email]

cc

Objet

Ca-measurement in glandular tissue

 






Dear List members,

we would like measure Ca(2+) in glandular tissue with Fluo3 using
confocal mircoscopy. We tried to load the tissue pieces with different
concentrations of the dye for 1,5 hours in 37C, washing after that
with PBS, and the tissue was put in a perfusion system. But we could
not see any changes in the fluorescence intensity when giving either
ATP or A23187 ionophore to the perfusion system.

Do you have any experience what would be an optimal loading protocol
or Fluo3 is suitable for this kind of measurements?

Thanks is advence!

Best regards,

Attila

--
Attila Cselenyak
PhD student

Semmelweis University
Basic Medical Science Center
Institute Of Human Physiology And Clinical Experimental Research
H-1094, Budapest
Tűzoltó utca 37-47.
Phone: +36-1-4591500 ext.60342
Mobile: +36-70-3874931