Alison Dun |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hiya I am looking for a source for DNA Origami using fluorescent dyes, does anyone know if I can get these commercially? Alternatively anything of a known structure/size that I can use for calibration of super-resolution microscopy - so between 50-100 nm in size. Thanks in advance Ali Dr. Alison Dun Technology and Facility Manager Edinburgh Super Resolution Imaging Consortium (ESRIC) Heriot-Watt University Edinburgh EH14 4AS Deliveries: ℅ Mary Pratt, EPS Stores Office: 3rd floor William Perkin Annexe Lab: 3.17 William Perkin Building Lab Phone: 0131 451 4748 Mobile: +44 7977 518 581 Alternative email: [hidden email] http://www.esric.org @AliDun_esric Register now for the ESRIC Summer School at http://www.esric.org/summer-school.html |
![]() ![]() |
John Oreopoulos |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Alison, I recently became aware of a small company that does just that. Here is the link (no commercial interest): http://www.gattaquant.com/ Haven't tried them yet though. Sincerely, John Oreopoulos Staff Scientist Spectral Applied Research Inc. A Division of Andor Technology Richmond Hill, Ontario Canada www.spectral.ca On 2015-03-24, at 9:52 AM, Alison Dun wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hiya > > I am looking for a source for DNA Origami using fluorescent dyes, does anyone know if I can get these commercially? Alternatively anything of a known structure/size that I can use for calibration of super-resolution microscopy - so between 50-100 nm in size. > > Thanks in advance > > Ali > > Dr. Alison Dun > Technology and Facility Manager > Edinburgh Super Resolution Imaging Consortium (ESRIC) > Heriot-Watt University > Edinburgh > EH14 4AS > > Deliveries: ℅ Mary Pratt, EPS Stores > Office: 3rd floor William Perkin Annexe > Lab: 3.17 William Perkin Building > Lab Phone: 0131 451 4748 > Mobile: +44 7977 518 581 > Alternative email: [hidden email] > > http://www.esric.org > @AliDun_esric > > Register now for the ESRIC Summer School at http://www.esric.org/summer-school.html |
![]() ![]() |
Seamus Holden |
![]() |
In reply to this post by Alison Dun
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Alison Philip Tinnefeld has a spin-out company (no commercial interest) selling DNA-origami nano rulers for super-res calibration. I've never tried them (just heard one of his students give a talk about it), but it looks like they do exactly what you are looking for. http://www.gattaquant.com/ http://www.gattaquant.com/products/localization-based.html Best wishes Seamus -- Dr Seamus Holden Post-doctoral Fellow Bâtiment des Sciences Physiques École Polytechnique Fédérale de Lausanne CH-1015 Lausanne, Switzerland Phone: +41-(0)21-693-0557 On 24.03.2015 14:52, Alison Dun wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hiya > > I am looking for a source for DNA Origami using fluorescent dyes, does anyone know if I can get these commercially? Alternatively anything of a known structure/size that I can use for calibration of super-resolution microscopy - so between 50-100 nm in size. > > Thanks in advance > > Ali > > Dr. Alison Dun > Technology and Facility Manager > Edinburgh Super Resolution Imaging Consortium (ESRIC) > Heriot-Watt University > Edinburgh > EH14 4AS > > Deliveries: ℅ Mary Pratt, EPS Stores > Office: 3rd floor William Perkin Annexe > Lab: 3.17 William Perkin Building > Lab Phone: 0131 451 4748 > Mobile: +44 7977 518 581 > Alternative email: [hidden email] > > http://www.esric.org > @AliDun_esric > > Register now for the ESRIC Summer School at http://www.esric.org/summer-school.html |
![]() ![]() |
Alison Dun |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Thanks Seamus, these look great! Ali Dr. Alison Dun Technology and Facility Manager Edinburgh Super Resolution Imaging Consortium (ESRIC) Heriot-Watt University Edinburgh EH14 4AS Deliveries: ℅ Mary Pratt, EPS Stores Office: 3rd floor William Perkin Annexe Lab: 3.17 William Perkin Building Lab Phone: 0131 451 4748 Mobile: +44 7977 518 581 Alternative email: [hidden email] http://www.esric.org @AliDun_esric Register now for the ESRIC Summer School at http://www.esric.org/summer-school.html > On 24 Mar 2015, at 14:38, Seamus Holden <[hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear Alison > > Philip Tinnefeld has a spin-out company (no commercial interest) selling DNA-origami nano rulers for super-res calibration. I've never tried them (just heard one of his students give a talk about it), but it looks like they do exactly what you are looking for. > > http://www.gattaquant.com/ > http://www.gattaquant.com/products/localization-based.html > > Best wishes > Seamus > > -- > Dr Seamus Holden > Post-doctoral Fellow > Bâtiment des Sciences Physiques > École Polytechnique Fédérale de Lausanne > CH-1015 Lausanne, Switzerland > > Phone: +41-(0)21-693-0557 > > > > On 24.03.2015 14:52, Alison Dun wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Hiya >> >> I am looking for a source for DNA Origami using fluorescent dyes, does anyone know if I can get these commercially? Alternatively anything of a known structure/size that I can use for calibration of super-resolution microscopy - so between 50-100 nm in size. >> >> Thanks in advance >> >> Ali >> >> Dr. Alison Dun >> Technology and Facility Manager >> Edinburgh Super Resolution Imaging Consortium (ESRIC) >> Heriot-Watt University >> Edinburgh >> EH14 4AS >> >> Deliveries: ℅ Mary Pratt, EPS Stores >> Office: 3rd floor William Perkin Annexe >> Lab: 3.17 William Perkin Building >> Lab Phone: 0131 451 4748 >> Mobile: +44 7977 518 581 >> Alternative email: [hidden email] >> >> http://www.esric.org >> @AliDun_esric >> >> Register now for the ESRIC Summer School at http://www.esric.org/summer-school.html |
![]() ![]() |
Alison J. North |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Ali, My 80 nm Green/Red PAINT slide from Gattaquant literally arrived last night, and we will be testing it out on our STORM system this afternoon. So feel free to get in touch and ask how we like it after today.... Best, Alison On 3/24/2015 10:56 AM, Alison Dun wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Thanks Seamus, these look great! > > Ali > > Dr. Alison Dun > Technology and Facility Manager > Edinburgh Super Resolution Imaging Consortium (ESRIC) > Heriot-Watt University > Edinburgh > EH14 4AS > > Deliveries: ℅ Mary Pratt, EPS Stores > Office: 3rd floor William Perkin Annexe > Lab: 3.17 William Perkin Building > Lab Phone: 0131 451 4748 > Mobile: +44 7977 518 581 > Alternative email: [hidden email] > > http://www.esric.org > @AliDun_esric > > Register now for the ESRIC Summer School at http://www.esric.org/summer-school.html > > > > > > >> On 24 Mar 2015, at 14:38, Seamus Holden <[hidden email]> wrote: >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Dear Alison >> >> Philip Tinnefeld has a spin-out company (no commercial interest) selling DNA-origami nano rulers for super-res calibration. I've never tried them (just heard one of his students give a talk about it), but it looks like they do exactly what you are looking for. >> >> http://www.gattaquant.com/ >> http://www.gattaquant.com/products/localization-based.html >> >> Best wishes >> Seamus >> >> -- >> Dr Seamus Holden >> Post-doctoral Fellow >> Bâtiment des Sciences Physiques >> École Polytechnique Fédérale de Lausanne >> CH-1015 Lausanne, Switzerland >> >> Phone: +41-(0)21-693-0557 >> >> >> >> On 24.03.2015 14:52, Alison Dun wrote: >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> Post images on http://www.imgur.com and include the link in your posting. >>> ***** >>> >>> Hiya >>> >>> I am looking for a source for DNA Origami using fluorescent dyes, does anyone know if I can get these commercially? Alternatively anything of a known structure/size that I can use for calibration of super-resolution microscopy - so between 50-100 nm in size. >>> >>> Thanks in advance >>> >>> Ali >>> >>> Dr. Alison Dun >>> Technology and Facility Manager >>> Edinburgh Super Resolution Imaging Consortium (ESRIC) >>> Heriot-Watt University >>> Edinburgh >>> EH14 4AS >>> >>> Deliveries: ℅ Mary Pratt, EPS Stores >>> Office: 3rd floor William Perkin Annexe >>> Lab: 3.17 William Perkin Building >>> Lab Phone: 0131 451 4748 >>> Mobile: +44 7977 518 581 >>> Alternative email: [hidden email] >>> >>> http://www.esric.org >>> @AliDun_esric >>> >>> Register now for the ESRIC Summer School at http://www.esric.org/summer-school.html -- Alison J. North, Ph.D., Senior Director of the Bio-Imaging Resource Center and Research Associate Professor, The Rockefeller University, 1230 York Avenue, New York, NY 10065. Tel: office ++ 212 327 7488 Tel: lab ++ 212 327 7486 Fax: ++ 212 327 7489 |
![]() ![]() |
Alison Dun |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Thanks everyone had so many responses from you all and overwhelming answer is gattaquant, will definitely be giving them a try and Alison yes I will do! Cheers Ali Dr. Alison Dun Technology and Facility Manager Edinburgh Super Resolution Imaging Consortium (ESRIC) Heriot-Watt University Edinburgh EH14 4AS Deliveries: ℅ Mary Pratt, EPS Stores Office: 3rd floor William Perkin Annexe Lab: 3.17 William Perkin Building Lab Phone: 0131 451 4748 Mobile: +44 7977 518 581 Alternative email: [hidden email] http://www.esric.org @AliDun_esric Register now for the ESRIC Summer School at http://www.esric.org/summer-school.html > On 24 Mar 2015, at 15:59, Alison Dun <[hidden email]> wrote: > > Hi > > Great stuff, I have had so many emails suggesting this company, I have requested a quote and yes I will do.. > > Thank you! > > Ali > > Dr. Alison Dun > Technology and Facility Manager > Edinburgh Super Resolution Imaging Consortium (ESRIC) > Heriot-Watt University > Edinburgh > EH14 4AS > > Deliveries: ℅ Mary Pratt, EPS Stores > Office: 3rd floor William Perkin Annexe > Lab: 3.17 William Perkin Building > Lab Phone: 0131 451 4748 > Mobile: +44 7977 518 581 > Alternative email: [hidden email] > > http://www.esric.org > @AliDun_esric > > Register now for the ESRIC Summer School at http://www.esric.org/summer-school.html > > > > > > >> On 24 Mar 2015, at 15:06, Alison North <[hidden email]> wrote: >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Hi Ali, >> >> My 80 nm Green/Red PAINT slide from Gattaquant literally arrived last night, and we will be testing it out on our STORM system this afternoon. So feel free to get in touch and ask how we like it after today.... >> >> Best, >> Alison >> >> >> On 3/24/2015 10:56 AM, Alison Dun wrote: >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> Post images on http://www.imgur.com and include the link in your posting. >>> ***** >>> >>> Thanks Seamus, these look great! >>> >>> Ali >>> >>> Dr. Alison Dun >>> Technology and Facility Manager >>> Edinburgh Super Resolution Imaging Consortium (ESRIC) >>> Heriot-Watt University >>> Edinburgh >>> EH14 4AS >>> >>> Deliveries: ℅ Mary Pratt, EPS Stores >>> Office: 3rd floor William Perkin Annexe >>> Lab: 3.17 William Perkin Building >>> Lab Phone: 0131 451 4748 >>> Mobile: +44 7977 518 581 >>> Alternative email: [hidden email] >>> >>> http://www.esric.org >>> @AliDun_esric >>> >>> Register now for the ESRIC Summer School at http://www.esric.org/summer-school.html >>> >>> >>> >>> >>> >>> >>>> On 24 Mar 2015, at 14:38, Seamus Holden <[hidden email]> wrote: >>>> >>>> ***** >>>> To join, leave or search the confocal microscopy listserv, go to: >>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>> Post images on http://www.imgur.com and include the link in your posting. >>>> ***** >>>> >>>> Dear Alison >>>> >>>> Philip Tinnefeld has a spin-out company (no commercial interest) selling DNA-origami nano rulers for super-res calibration. I've never tried them (just heard one of his students give a talk about it), but it looks like they do exactly what you are looking for. >>>> >>>> http://www.gattaquant.com/ >>>> http://www.gattaquant.com/products/localization-based.html >>>> >>>> Best wishes >>>> Seamus >>>> >>>> -- >>>> Dr Seamus Holden >>>> Post-doctoral Fellow >>>> Bâtiment des Sciences Physiques >>>> École Polytechnique Fédérale de Lausanne >>>> CH-1015 Lausanne, Switzerland >>>> >>>> Phone: +41-(0)21-693-0557 >>>> >>>> >>>> >>>> On 24.03.2015 14:52, Alison Dun wrote: >>>>> ***** >>>>> To join, leave or search the confocal microscopy listserv, go to: >>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>> Post images on http://www.imgur.com and include the link in your posting. >>>>> ***** >>>>> >>>>> Hiya >>>>> >>>>> I am looking for a source for DNA Origami using fluorescent dyes, does anyone know if I can get these commercially? Alternatively anything of a known structure/size that I can use for calibration of super-resolution microscopy - so between 50-100 nm in size. >>>>> >>>>> Thanks in advance >>>>> >>>>> Ali >>>>> >>>>> Dr. Alison Dun >>>>> Technology and Facility Manager >>>>> Edinburgh Super Resolution Imaging Consortium (ESRIC) >>>>> Heriot-Watt University >>>>> Edinburgh >>>>> EH14 4AS >>>>> >>>>> Deliveries: ℅ Mary Pratt, EPS Stores >>>>> Office: 3rd floor William Perkin Annexe >>>>> Lab: 3.17 William Perkin Building >>>>> Lab Phone: 0131 451 4748 >>>>> Mobile: +44 7977 518 581 >>>>> Alternative email: [hidden email] >>>>> >>>>> http://www.esric.org >>>>> @AliDun_esric >>>>> >>>>> Register now for the ESRIC Summer School at http://www.esric.org/summer-school.html >> >> -- >> Alison J. North, Ph.D., >> Senior Director of the Bio-Imaging Resource Center and >> Research Associate Professor, >> The Rockefeller University, >> 1230 York Avenue, >> New York, >> NY 10065. >> Tel: office ++ 212 327 7488 >> Tel: lab ++ 212 327 7486 >> Fax: ++ 212 327 7489 > |
![]() ![]() |
Christian Soeller |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Again a bump for gattaquant. These definitely work well, I have had one for about two months, but you may have to use a modified filtering algorithm for best results (reject diffusive backgrounds in PAINT). Cheers, Christian -- Christian Soeller PhD - Dept. of Physiology - Univ. of Auckland Sent with Sparrow (http://www.sparrowmailapp.com/?sig) On Tuesday, 24 March 2015 at 4:01 PM, Alison Dun wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Thanks everyone had so many responses from you all and overwhelming answer is gattaquant, will definitely be giving them a try and Alison yes I will do! > > Cheers > Ali > > Dr. Alison Dun > Technology and Facility Manager > Edinburgh Super Resolution Imaging Consortium (ESRIC) > Heriot-Watt University > Edinburgh > EH14 4AS > > Deliveries: ℅ Mary Pratt, EPS Stores > Office: 3rd floor William Perkin Annexe > Lab: 3.17 William Perkin Building > Lab Phone: 0131 451 4748 > Mobile: +44 7977 518 581 > Alternative email: [hidden email] (mailto:[hidden email]) > > http://www.esric.org > @AliDun_esric > > Register now for the ESRIC Summer School at http://www.esric.org/summer-school.html > > > > > > > > On 24 Mar 2015, at 15:59, Alison Dun <[hidden email] (mailto:[hidden email])> wrote: > > > > Hi > > > > Great stuff, I have had so many emails suggesting this company, I have requested a quote and yes I will do.. > > > > Thank you! > > > > Ali > > > > Dr. Alison Dun > > Technology and Facility Manager > > Edinburgh Super Resolution Imaging Consortium (ESRIC) > > Heriot-Watt University > > Edinburgh > > EH14 4AS > > > > Deliveries: ℅ Mary Pratt, EPS Stores > > Office: 3rd floor William Perkin Annexe > > Lab: 3.17 William Perkin Building > > Lab Phone: 0131 451 4748 > > Mobile: +44 7977 518 581 > > Alternative email: [hidden email] (mailto:[hidden email]) > > > > http://www.esric.org > > @AliDun_esric > > > > Register now for the ESRIC Summer School at http://www.esric.org/summer-school.html > > > > > > > > > > > > > > > On 24 Mar 2015, at 15:06, Alison North <[hidden email] (mailto:[hidden email])> wrote: > > > > > > ***** > > > To join, leave or search the confocal microscopy listserv, go to: > > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > > Post images on http://www.imgur.com and include the link in your posting. > > > ***** > > > > > > Hi Ali, > > > > > > My 80 nm Green/Red PAINT slide from Gattaquant literally arrived last night, and we will be testing it out on our STORM system this afternoon. So feel free to get in touch and ask how we like it after today.... > > > > > > Best, > > > Alison > > > > > > > > > On 3/24/2015 10:56 AM, Alison Dun wrote: > > > > ***** > > > > To join, leave or search the confocal microscopy listserv, go to: > > > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > > > Post images on http://www.imgur.com and include the link in your posting. > > > > ***** > > > > > > > > Thanks Seamus, these look great! > > > > > > > > Ali > > > > > > > > Dr. Alison Dun > > > > Technology and Facility Manager > > > > Edinburgh Super Resolution Imaging Consortium (ESRIC) > > > > Heriot-Watt University > > > > Edinburgh > > > > EH14 4AS > > > > > > > > Deliveries: ℅ Mary Pratt, EPS Stores > > > > Office: 3rd floor William Perkin Annexe > > > > Lab: 3.17 William Perkin Building > > > > Lab Phone: 0131 451 4748 > > > > Mobile: +44 7977 518 581 > > > > Alternative email: [hidden email] (mailto:[hidden email]) > > > > > > > > http://www.esric.org > > > > @AliDun_esric > > > > > > > > Register now for the ESRIC Summer School at http://www.esric.org/summer-school.html > > > > > > > > > > > > > > > > > > > > > > > > > > > > > On 24 Mar 2015, at 14:38, Seamus Holden <[hidden email] (mailto:[hidden email])> wrote: > > > > > > > > > > ***** > > > > > To join, leave or search the confocal microscopy listserv, go to: > > > > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > > > > Post images on http://www.imgur.com and include the link in your posting. > > > > > ***** > > > > > > > > > > Dear Alison > > > > > > > > > > Philip Tinnefeld has a spin-out company (no commercial interest) selling DNA-origami nano rulers for super-res calibration. I've never tried them (just heard one of his students give a talk about it), but it looks like they do exactly what you are looking for. > > > > > > > > > > http://www.gattaquant.com/ > > > > > http://www.gattaquant.com/products/localization-based.html > > > > > > > > > > Best wishes > > > > > Seamus > > > > > > > > > > -- > > > > > Dr Seamus Holden > > > > > Post-doctoral Fellow > > > > > Bâtiment des Sciences Physiques > > > > > École Polytechnique Fédérale de Lausanne > > > > > CH-1015 Lausanne, Switzerland > > > > > > > > > > Phone: +41-(0)21-693-0557 > > > > > > > > > > > > > > > > > > > > On 24.03.2015 14:52, Alison Dun wrote: > > > > > > ***** > > > > > > To join, leave or search the confocal microscopy listserv, go to: > > > > > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > > > > > Post images on http://www.imgur.com and include the link in your posting. > > > > > > ***** > > > > > > > > > > > > Hiya > > > > > > > > > > > > I am looking for a source for DNA Origami using fluorescent dyes, does anyone know if I can get these commercially? Alternatively anything of a known structure/size that I can use for calibration of super-resolution microscopy - so between 50-100 nm in size. > > > > > > > > > > > > Thanks in advance > > > > > > > > > > > > Ali > > > > > > > > > > > > Dr. Alison Dun > > > > > > Technology and Facility Manager > > > > > > Edinburgh Super Resolution Imaging Consortium (ESRIC) > > > > > > Heriot-Watt University > > > > > > Edinburgh > > > > > > EH14 4AS > > > > > > > > > > > > Deliveries: ℅ Mary Pratt, EPS Stores > > > > > > Office: 3rd floor William Perkin Annexe > > > > > > Lab: 3.17 William Perkin Building > > > > > > Lab Phone: 0131 451 4748 > > > > > > Mobile: +44 7977 518 581 > > > > > > Alternative email: [hidden email] (mailto:[hidden email]) > > > > > > > > > > > > http://www.esric.org > > > > > > @AliDun_esric > > > > > > > > > > > > Register now for the ESRIC Summer School at http://www.esric.org/summer-school.html > > > > > > -- > > > Alison J. North, Ph.D., > > > Senior Director of the Bio-Imaging Resource Center and > > > Research Associate Professor, > > > The Rockefeller University, > > > 1230 York Avenue, > > > New York, > > > NY 10065. > > > Tel: office ++ 212 327 7488 > > > Tel: lab ++ 212 327 7486 > > > Fax: ++ 212 327 7489 > > > > > > > > > > |