Dear All,

We are considering to purchase (write a proposal) for a light sheet system. For the completeness of the provider list I wanted to ask you whether you  are aware of any other commercial system then the one from: Zeiss Z1), Leica (DLS), Luxendo, I3 (based on the ASI setup) and LaVision (Ultramicroscope).
        I know that there are considerable differences between the application fields (sample mounting etc.) of these different systems, but wanted to make sure that I have the complete list. Furthermore, is there anyone who have compared the Leica solution to other ones? I would very much appreciate if you could share your feedback either on the list or privately.

Thanks Gabor
Philippe clemenceau Philippe clemenceau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Commercial light sheet systems

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hello Gabor ,


The ALPHA3 system from Phaseview ( France) is another option for light sheet microscopy recently launched on the market but with some clear distinctions from other existing systems.

You should check the following link : http://www.alphalightsheet.com

We are the distributor in the US and Canada, for other territories, please contact Phaseview SA directly !

Regards,

Philippe Clémenceau,
Co-Founder, MS in Optical Science
Axiom Optics, Excellence in advanced imaging and optical metrology
 
14 Tyler St, 3rd floor
Somerville, MA 02143, USA
 
Phone:  (617) 401 2198
Mobile: (617) 869 3131

Check out our product line at www.axiomoptics.com <http://www.axiomoptics.com/>  


> On Feb 24, 2017, at 5:40 AM, Csúcs Gábor <[hidden email]> wrote:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
>
> Dear All,
>
> We are considering to purchase (write a proposal) for a light sheet system. For the completeness of the provider list I wanted to ask you whether you  are aware of any other commercial system then the one from: Zeiss Z1), Leica (DLS), Luxendo, I3 (based on the ASI setup) and LaVision (Ultramicroscope).
> I know that there are considerable differences between the application fields (sample mounting etc.) of these different systems, but wanted to make sure that I have the complete list. Furthermore, is there anyone who have compared the Leica solution to other ones? I would very much appreciate if you could share your feedback either on the list or privately.
>
> Thanks Gabor
Weber, Michael Weber, Michael
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Commercial light sheet systems

In reply to this post by Csúcs Gábor-3
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Gabor,

Here’s my list of commercial light sheet systems that I’m aware of. There are a few that you didn’t mention.

- Zeiss Lightsheet Z1
- Leica SP8 DLS
- ASI diSPIM
- ASI oSPIM
- 3I Marianas
- 3I Lattice Lightsheet
- LaVision Ultramicroscope
- Cairn L-SPI Illuminator
- Luxendo MuVi-SPIM
- Luxendo InVi-SPIM
- Phaseview Alpha3

Those are the future setups that I know of (in case you have time to wait):

- Zeiss Lattice
- Leica SCAPE

I got a first impression of the Leica DLS last year, and it’s definitely interesting. You need a full confocal to get the light sheet illumination, which might be a pro or con. Personally, I’d like to have both techniques in separate setups, as they serve different purposes. It would be nice to have a simplified, matched scanhead without detectors, just for light sheet illumination. One could still use it for sequential photo-manipulation and light sheet imaging.

Best,
Michael

On Feb 24, 2017, at 5:40 AM, Csúcs Gábor <[hidden email]<mailto:[hidden email]>> wrote:

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****


Dear All,

We are considering to purchase (write a proposal) for a light sheet system. For the completeness of the provider list I wanted to ask you whether you  are aware of any other commercial system then the one from: Zeiss Z1), Leica (DLS), Luxendo, I3 (based on the ASI setup) and LaVision (Ultramicroscope).
I know that there are considerable differences between the application fields (sample mounting etc.) of these different systems, but wanted to make sure that I have the complete list. Furthermore, is there anyone who have compared the Leica solution to other ones? I would very much appreciate if you could share your feedback either on the list or privately.

Thanks Gabor

_____________

Dr. Michael Weber
Advanced Microscopy Fellow
Harvard Medical School
240 Longwood Ave, LHRRB 113, Boston, MA 02115
http://nic.med.harvard.edu

phil laissue-2 phil laissue-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Commercial light sheet systems

In reply to this post by Philippe clemenceau
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Gabor,

Cairn Research (UK) produce the L-SPI (L-shaped single plane illuminator).
It's completely platform-independent and can be adapted to diverse
experimental settings. It's based on a prototype I made for looking at
corals, so it works well with large, photosensitive samples. Papers on the
setup and applications should be published soon. Until then, please feel
free to contact me offline for more information.

With kind regards

Philippe Laissue

Note: I work with Cairn Research in the context of my industrial
fellowship, but I have no financial interest in the L-SPI. I'm just jazzed
to see it get used:)



_________________________________________
Philippe Laissue, PhD
Royal Society Industry Fellow
School of Biological Sciences, Room 4.17
University of Essex, Colchester CO4 3SQ, UK
(0044) 01206 872246 / (0044) 07842 676 456
[hidden email]
privatewww.essex.ac.uk/~plaissue

On 24 February 2017 at 15:36, Philippe Clemenceau <
[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hello Gabor ,
>
>
> The ALPHA3 system from Phaseview ( France) is another option for light
> sheet microscopy recently launched on the market but with some clear
> distinctions from other existing systems.
>
> You should check the following link : http://www.alphalightsheet.com
>
> We are the distributor in the US and Canada, for other territories, please
> contact Phaseview SA directly !
>
> Regards,
>
> Philippe Clémenceau,
> Co-Founder, MS in Optical Science
> Axiom Optics, Excellence in advanced imaging and optical metrology
>
> 14 Tyler St, 3rd floor
> Somerville, MA 02143, USA
>
> Phone:  (617) 401 2198
> Mobile: (617) 869 3131
>
> Check out our product line at www.axiomoptics.com <
> http://www.axiomoptics.com/>
>
>
> > On Feb 24, 2017, at 5:40 AM, Csúcs Gábor <[hidden email]>
> wrote:
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > Post images on http://www.imgur.com and include the link in your
> posting.
> > *****
> >
> >
> > Dear All,
> >
> > We are considering to purchase (write a proposal) for a light sheet
> system. For the completeness of the provider list I wanted to ask you
> whether you  are aware of any other commercial system then the one from:
> Zeiss Z1), Leica (DLS), Luxendo, I3 (based on the ASI setup) and LaVision
> (Ultramicroscope).
> >       I know that there are considerable differences between the
> application fields (sample mounting etc.) of these different systems, but
> wanted to make sure that I have the complete list. Furthermore, is there
> anyone who have compared the Leica solution to other ones? I would very
> much appreciate if you could share your feedback either on the list or
> privately.
> >
> > Thanks                        Gabor
>
Jim Mansfield Jim Mansfield
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Commercial light sheet systems

In reply to this post by Weber, Michael
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

A very good list!

Just to add to it, I believe that M2 Lasers (M-Squared Lasers) also has an Airy light sheet. They don't seem to have it on their website, but they were publicizing it at BiOS.

https://www.photonics.com/Article.aspx?AID=60716

Jim

(no commercial interest)


Jim Mansfield
[hidden email]
+1-204-619-3350

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Weber, Michael
Sent: February 24, 2017 9:54 AM
To: [hidden email]
Subject: Re: Commercial light sheet systems

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Gabor,

Here’s my list of commercial light sheet systems that I’m aware of. There are a few that you didn’t mention.

- Zeiss Lightsheet Z1
- Leica SP8 DLS
- ASI diSPIM
- ASI oSPIM
- 3I Marianas
- 3I Lattice Lightsheet
- LaVision Ultramicroscope
- Cairn L-SPI Illuminator
- Luxendo MuVi-SPIM
- Luxendo InVi-SPIM
- Phaseview Alpha3

Those are the future setups that I know of (in case you have time to wait):

- Zeiss Lattice
- Leica SCAPE

I got a first impression of the Leica DLS last year, and it’s definitely interesting. You need a full confocal to get the light sheet illumination, which might be a pro or con. Personally, I’d like to have both techniques in separate setups, as they serve different purposes. It would be nice to have a simplified, matched scanhead without detectors, just for light sheet illumination. One could still use it for sequential photo-manipulation and light sheet imaging.

Best,
Michael

On Feb 24, 2017, at 5:40 AM, Csúcs Gábor <[hidden email]<mailto:[hidden email]>> wrote:

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****


Dear All,

We are considering to purchase (write a proposal) for a light sheet system. For the completeness of the provider list I wanted to ask you whether you  are aware of any other commercial system then the one from: Zeiss Z1), Leica (DLS), Luxendo, I3 (based on the ASI setup) and LaVision (Ultramicroscope).
I know that there are considerable differences between the application fields (sample mounting etc.) of these different systems, but wanted to make sure that I have the complete list. Furthermore, is there anyone who have compared the Leica solution to other ones? I would very much appreciate if you could share your feedback either on the list or privately.

Thanks Gabor

_____________

Dr. Michael Weber
Advanced Microscopy Fellow
Harvard Medical School
240 Longwood Ave, LHRRB 113, Boston, MA 02115 http://nic.med.harvard.edu
darren.thomson darren.thomson
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Commercial light sheet systems

In reply to this post by Csúcs Gábor-3
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

We've used the DLS and the Z1 to look at agar enclosed spores and cleared lung samples. The DLS needs to have your sample mounted on a small dish, whereas the Z1 needs you to suspend your sample into their enclosed chamber. They have capillary tubes, butcher hooks. With the Z1 there is scope to 3D print your own stage holders. Something to think about.

With the DLS you have the potential to combine LSFM with traditional confocal microscopy, so for example you could bleach/ablate a region with the confocal set up, then switch to LSFM to monitor the whole system.
I think the DLS comes in at ~£70 000 as an add on to your confocal, whereas the Z1 may be ~£300 000 I think...(please correct if I'm wrong)

They are the only systems I've had experience with.

Please account for significant investment in the PC and storage side of things. These things spew out data at an alarming rate if you're not prepared.

Happy hunting!

Darren

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Csúcs Gábor
Sent: 24 February 2017 10:41
To: [hidden email]
Subject: Commercial light sheet systems

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****


Dear All,

We are considering to purchase (write a proposal) for a light sheet system. For the completeness of the provider list I wanted to ask you whether you  are aware of any other commercial system then the one from: Zeiss Z1), Leica (DLS), Luxendo, I3 (based on the ASI setup) and LaVision (Ultramicroscope).
        I know that there are considerable differences between the application fields (sample mounting etc.) of these different systems, but wanted to make sure that I have the complete list. Furthermore, is there anyone who have compared the Leica solution to other ones? I would very much appreciate if you could share your feedback either on the list or privately.

Thanks Gabor
Csúcs  Gábor-3 Csúcs Gábor-3
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Commercial light sheet systems

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear All,

Thanks for all of you who shared information about the light sheet systems! I learned again a lot, this list is an amaring resource.

Gabor