Controlled DNA damage by lasers

classic Classic list List threaded Threaded
♦
♦
10 messages Options
Michal Jarnik Michal Jarnik
Reply | Threaded
Open this post in threaded view
♦
♦
|

Controlled DNA damage by lasers

Listers,

We would need to perform experiments including controlled DNA damage to
cells in culture. One idea was to use the LSCM and expose an area in a
way similar to the bleaching part of a FRAP experiment. We do have a LED
405 nm laser (I would assume that would be the right wavelength, we do
not have any really UV laser), Nikon C1 can do quite a lot of zooming to
concentrate the energy in a small area. I have no idea, though, about
the power necessary/available and other conditions. Any recommendations
would be very much appreciated.

Thanks, M.

--

Michal Jarnik, Ph.D.
Cell Imaging Facility
Electron Microscope Facility
Fox Chase Cancer Center
333 Cottman Ave.
Philadelphia, PA 19111
215-728-5675 (v.)
215-728-2770 (v.)
215-728-2412 (f.)
Mario-2 Mario-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Controlled DNA damage by lasers

Michal,

It would help to know what you mean by controlled DNA damage. #
Strand breaks, repair errors, cell killing? A 405 nm laser is by
itself not particularly specific for DNA whose bases absorb much more
strongly from 240-280 nm and not much at all above 320 nm (as far as
I remember). However, dye molecules located near or bound to DNA can
absorb longer wavelengths generating photo-toxic products such as
singlet oxygen, superoxide, *OH, phenoxyls, etc. You might be able to
create a dye/pigment enhanced system using the 405 nm laser.
Multiphoton excitation might work, e.g., three photon absorption ~800
nm. But that means an entirely different sort of system and the
chances of frying your cells becomes very probable.

As for zooming, you must mean this is in the most generic sense since
for any given objective its numerical aperture will always create the
same energy density profile given the same amount of illumination,
unless you mean to move it out of focus on the target. I am not sure
that this is the best way to provide even illumination. Anyway, more
details could be helpful.

Mario


>Listers,
>
>We would need to perform experiments including controlled DNA damage
>to cells in culture. One idea was to use the LSCM and expose an area
>in a way similar to the bleaching part of a FRAP experiment. We do
>have a LED 405 nm laser (I would assume that would be the right
>wavelength, we do not have any really UV laser), Nikon C1 can do
>quite a lot of zooming to concentrate the energy in a small area. I
>have no idea, though, about the power necessary/available and other
>conditions. Any recommendations would be very much appreciated.
>
>Thanks, M.
>
>--
>
>Michal Jarnik, Ph.D.
>Cell Imaging Facility
>Electron Microscope Facility
>Fox Chase Cancer Center
>333 Cottman Ave.
>Philadelphia, PA 19111
>215-728-5675 (v.)
>215-728-2770 (v.)
>215-728-2412 (f.)


--
________________________________________________________________________________
Mario M. Moronne, Ph.D.

[hidden email]
[hidden email]
Tej Hiran Tej Hiran
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Controlled DNA damage by lasers

In reply to this post by Michal Jarnik
Hi Michal,

I didn't really follow the entire link but I induce DNA damage to cells for Tunnel analysis by flow cytometry by simply exposing our cells to the UV lamp of the tissue culture hood.  Normally 30 minutes is sufficient and after 48 hours our cells (A549) show very high levels of apoptosis.  The only drawback it that it is not very controlled - other than time.  Hopefully this helps.

Tej.

> Date: Mon, 21 Jun 2010 12:20:17 -0400
> From: [hidden email]
> Subject: Controlled DNA damage by lasers
> To: [hidden email]
>
> Listers,
>
> We would need to perform experiments including controlled DNA damage to
> cells in culture. One idea was to use the LSCM and expose an area in a
> way similar to the bleaching part of a FRAP experiment. We do have a LED
> 405 nm laser (I would assume that would be the right wavelength, we do
> not have any really UV laser), Nikon C1 can do quite a lot of zooming to
> concentrate the energy in a small area. I have no idea, though, about
> the power necessary/available and other conditions. Any recommendations
> would be very much appreciated.
>
> Thanks, M.
>
> --
>
> Michal Jarnik, Ph.D.
> Cell Imaging Facility
> Electron Microscope Facility
> Fox Chase Cancer Center
> 333 Cottman Ave.
> Philadelphia, PA 19111
> 215-728-5675 (v.)
> 215-728-2770 (v.)
> 215-728-2412 (f.)


The New Busy think 9 to 5 is a cute idea. Combine multiple calendars with Hotmail. Get busy.
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Controlled DNA damage by lasers

I agree with the other posters.  405nm is not short enough; it's
really more 'violet' than UV.  You would be better off with Tej's
suggestion of a good arc lamp with appropriate UV pass filters.  Just
focus the lamp onto a section of your cells and let them 'bake' a
while.  You will need a good filter to only let the appropriate UV
band through, otherwise the excess energy at other wavelengths will do
other forms of damage to your cells.

Craig

On Mon, Jun 21, 2010 at 11:45 AM, Tej Hiran <[hidden email]> wrote:

> Hi Michal,
>
> I didn't really follow the entire link but I induce DNA damage to cells for
> Tunnel analysis by flow cytometry by simply exposing our cells to the UV
> lamp of the tissue culture hood.  Normally 30 minutes is sufficient and
> after 48 hours our cells (A549) show very high levels of apoptosis.  The
> only drawback it that it is not very controlled - other than time.
> Hopefully this helps.
>
> Tej.
>
>> Date: Mon, 21 Jun 2010 12:20:17 -0400
>> From: [hidden email]
>> Subject: Controlled DNA damage by lasers
>> To: [hidden email]
>>
>> Listers,
>>
>> We would need to perform experiments including controlled DNA damage to
>> cells in culture. One idea was to use the LSCM and expose an area in a
>> way similar to the bleaching part of a FRAP experiment. We do have a LED
>> 405 nm laser (I would assume that would be the right wavelength, we do
>> not have any really UV laser), Nikon C1 can do quite a lot of zooming to
>> concentrate the energy in a small area. I have no idea, though, about
>> the power necessary/available and other conditions. Any recommendations
>> would be very much appreciated.
>>
>> Thanks, M.
>>
>> --
>>
>> Michal Jarnik, Ph.D.
>> Cell Imaging Facility
>> Electron Microscope Facility
>> Fox Chase Cancer Center
>> 333 Cottman Ave.
>> Philadelphia, PA 19111
>> 215-728-5675 (v.)
>> 215-728-2770 (v.)
>> 215-728-2412 (f.)
>
> ________________________________
> The New Busy think 9 to 5 is a cute idea. Combine multiple calendars with
> Hotmail. Get busy.
Julio Vazquez Julio Vazquez
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Controlled DNA damage by lasers

In reply to this post by Michal Jarnik
Michal, 

DNA damage has been achieved with a 405 laser on a laser scanning confocal (such as in Bergink et al, Genes Dev 20: 1343-1352, 2006). People have also had success with a Photonic Instruments Micro-Point laser (Ayoub et al, Nature 453, 682-686, 2008). A FRAP of photokinesis laser module from a variety of vendors will also work fine. If you check the literature, you will see that different wavelengths may induce different types of DNA damage, also depending on which sensitizers you add to your cells, so doing a review of the literature would be a good idea. A PubMed search for " laser DNA damage" or similar will give you plenty to read. One review on different laser irradiation methods was published recently (Kong et al, NAR 37(9): e68-2009). Exact conditions will depend on your samples and experiment... as usual for this type of work, I would recommend using the most gentle settings that allow you to see a response.  

--
Julio Vazquez
Fred Hutchinson Cancer Research Center
Seattle, WA 98109-1024



On Jun 21, 2010, at 9:20 AM, Michal Jarnik wrote:

Listers,

We would need to perform experiments including controlled DNA damage to cells in culture. One idea was to use the LSCM and expose an area in a way similar to the bleaching part of a FRAP experiment. We do have a LED 405 nm laser (I would assume that would be the right wavelength, we do not have any really UV laser), Nikon C1 can do quite a lot of zooming to concentrate the energy in a small area. I have no idea, though, about the power necessary/available and other conditions. Any recommendations would be very much appreciated.

Thanks, M.

-- 

Michal Jarnik, Ph.D.
Cell Imaging Facility
Electron Microscope Facility
Fox Chase Cancer Center
333 Cottman Ave.
Philadelphia, PA 19111
215-728-5675 (v.)
215-728-2770 (v.)
215-728-2412 (f.)

Z.J. Zhang Z.J. Zhang
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Controlled DNA damage by lasers

In reply to this post by Tej Hiran

Hi Michal:

 

Did you consider using DNase? It would cause “DNA damage” in a more controlled way.

 

Zhaojie

 

 

Zhaojie Zhang, Ph. D.

Director, Microscopy Core Facility

University of Wyoming

Laramie, WY 82071

PHONE: 307-766-3038

FAX: 307-766-5625

 

 

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Tej Hiran
Sent: Monday, June 21, 2010 11:45 AM
To: [hidden email]
Subject: Re: Controlled DNA damage by lasers

 

Hi Michal,

I didn't really follow the entire link but I induce DNA damage to cells for Tunnel analysis by flow cytometry by simply exposing our cells to the UV lamp of the tissue culture hood.  Normally 30 minutes is sufficient and after 48 hours our cells (A549) show very high levels of apoptosis.  The only drawback it that it is not very controlled - other than time.  Hopefully this helps.

Tej.

> Date: Mon, 21 Jun 2010 12:20:17 -0400
> From: [hidden email]
> Subject: Controlled DNA damage by lasers
> To: [hidden email]
>
> Listers,
>
> We would need to perform experiments including controlled DNA damage to
> cells in culture. One idea was to use the LSCM and expose an area in a
> way similar to the bleaching part of a FRAP experiment. We do have a LED
> 405 nm laser (I would assume that would be the right wavelength, we do
> not have any really UV laser), Nikon C1 can do quite a lot of zooming to
> concentrate the energy in a small area. I have no idea, though, about
> the power necessary/available and other conditions. Any recommendations
> would be very much appreciated.
>
> Thanks, M.
>
> --
>
> Michal Jarnik, Ph.D.
> Cell Imaging Facility
> Electron Microscope Facility
> Fox Chase Cancer Center
> 333 Cottman Ave.
> Philadelphia, PA 19111
> 215-728-5675 (v.)
> 215-728-2770 (v.)
> 215-728-2412 (f.)


The New Busy think 9 to 5 is a cute idea. Combine multiple calendars with Hotmail. Get busy.

Adams,Henry P Adams,Henry P
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Controlled DNA damage by lasers

In reply to this post by Michal Jarnik
Michal,
We have a Micropoint targeted illumination setup from Photonic Instruments for inducing DNA damage. It uses a nitrogen 365nm laser that is directly coupled to the epifluorescence path on a Nikon TE2000 inverted microscope. The laser power, number of pulses and ROI are controlled by a Metamorph plug-in.  Investigators draw a line through the nucleus with the ROI tool and lase it. They usually lase 20 to 50 cells in one session, fix and stain. Or they use the timelapse feature when their proteins are coupled to FPs.
I have no stake in the company.

Hank Adams
Genetics Department
Rm S14.8316A
U.T. M.D.Anderson Cancer Center
Houston, Tx 77030

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Michal Jarnik
Sent: Monday, June 21, 2010 11:20 AM
To: [hidden email]
Subject: Controlled DNA damage by lasers

Listers,

We would need to perform experiments including controlled DNA damage to
cells in culture. One idea was to use the LSCM and expose an area in a
way similar to the bleaching part of a FRAP experiment. We do have a LED
405 nm laser (I would assume that would be the right wavelength, we do
not have any really UV laser), Nikon C1 can do quite a lot of zooming to
concentrate the energy in a small area. I have no idea, though, about
the power necessary/available and other conditions. Any recommendations
would be very much appreciated.

Thanks, M.

--

Michal Jarnik, Ph.D.
Cell Imaging Facility
Electron Microscope Facility
Fox Chase Cancer Center
333 Cottman Ave.
Philadelphia, PA 19111
215-728-5675 (v.)
215-728-2770 (v.)
215-728-2412 (f.)
Mario-2 Mario-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Controlled DNA damage by lasers

Re: Controlled DNA damage by lasers
Hank, Michal

Do you know the power per pulse and the lifetime of the Micropoint system (2-20 nsec is common for N2 lasers, I think)? As I recall, nitrogen lasers have a number of UV lines including 337 and 357 nm; the former might be particularly useful for Michal's DNA study. When I worked on creating lanthanide labels for x-ray microscopy, I considered using a 337 nm nitrogen laser as the excitor for part of a test platform. However, I developed an alternative especially as the nitrogen laser I was considering was particularly suited to ablation, and was sold specifically for that purpose.

This makes me think that unless one uses a sensitizer enabling longer wavelengths to be used (including 357 or 365 nm lines), it may be very difficult to produce anything that would be considered reproducible dose dependent DNA damage. What I imagine is DNA being surrounded by a poisonous soup of photolysis products. Even if you had quartz objectives enabling the use of selective Hg UV lines readily absorbed by DNA bases (e.g., 253 nm) or a noble gas flash lamp (1 usec. pulses), the surrounding nuclear proteins will also absorb the same light producing DNA damaging products in difficult to predict ways. Then again maybe this is what Michal wants.

Photodynamic therapy (PDT) has been around for nearly a century and there is a large body of literature pertaining to the use of sensitizers and long wavelength visible light. That would probably be a good place to look. Of newer approaches, soft x-ray illumination using the "water window" of the nucleus might be particularly interesting, as the cells can be targeted down from whole nuclei to very tiny sub volumes (< 50 nm x 50 nm x 10 um thick). This can be done with cells at cryogenic temperatures such that the primary damage would occur via direct targeting of DNA. Non DNA free radical damage would be minimized.

Mario


Michal,
We have a Micropoint targeted illumination setup from Photonic Instruments for inducing DNA damage. It uses a nitrogen 365nm laser that is directly coupled to the epifluorescence path on a Nikon TE2000 inverted microscope. The laser power, number of pulses and ROI are controlled by a Metamorph plug-in.  Investigators draw a line through the nucleus with the ROI tool and lase it. They usually lase 20 to 50 cells in one session, fix and stain. Or they use the timelapse feature when their proteins are coupled to FPs.
I have no stake in the company.

Hank Adams
Genetics Department
Rm S14.8316A
U.T. M.D.Anderson Cancer Center
Houston, Tx 77030

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Michal Jarnik
Sent: Monday, June 21, 2010 11:20 AM
To: [hidden email]
Subject: Controlled DNA damage by lasers

Listers,

We would need to perform experiments including controlled DNA damage to
cells in culture. One idea was to use the LSCM and expose an area in a
way similar to the bleaching part of a FRAP experiment. We do have a LED
405 nm laser (I would assume that would be the right wavelength, we do
not have any really UV laser), Nikon C1 can do quite a lot of zooming to
concentrate the energy in a small area. I have no idea, though, about
the power necessary/available and other conditions. Any recommendations
would be very much appreciated.

Thanks, M.

--

Michal Jarnik, Ph.D.
Cell Imaging Facility
Electron Microscope Facility
Fox Chase Cancer Center
333 Cottman Ave.
Philadelphia, PA 19111
215-728-5675 (v.)
215-728-2770 (v.)
215-728-2412 (f.)


-- 
________________________________________________________________________________
Mario M. Moronne, Ph.D.

[hidden email]
[hidden email]
Adams,Henry P Adams,Henry P
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Controlled DNA damage by lasers

Re: Controlled DNA damage by lasers

Mario,

It is actually a nitrogen laser-pumped dye laser. The nitrogen laser is ~120 ujoules of 337nm with a 2 to 6 nsec pulse length. The pulsed beam is directed through a low loss multimode fiber to a dye laser. The dye laser beam is 364nm and exits the objective.

Hank

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mario
Sent: Wednesday, June 23, 2010 12:21 PM
To: [hidden email]
Subject: Re: Controlled DNA damage by lasers

 

Hank, Michal

 

Do you know the power per pulse and the lifetime of the Micropoint system (2-20 nsec is common for N2 lasers, I think)? As I recall, nitrogen lasers have a number of UV lines including 337 and 357 nm; the former might be particularly useful for Michal's DNA study. When I worked on creating lanthanide labels for x-ray microscopy, I considered using a 337 nm nitrogen laser as the excitor for part of a test platform. However, I developed an alternative especially as the nitrogen laser I was considering was particularly suited to ablation, and was sold specifically for that purpose.

 

This makes me think that unless one uses a sensitizer enabling longer wavelengths to be used (including 357 or 365 nm lines), it may be very difficult to produce anything that would be considered reproducible dose dependent DNA damage. What I imagine is DNA being surrounded by a poisonous soup of photolysis products. Even if you had quartz objectives enabling the use of selective Hg UV lines readily absorbed by DNA bases (e.g., 253 nm) or a noble gas flash lamp (1 usec. pulses), the surrounding nuclear proteins will also absorb the same light producing DNA damaging products in difficult to predict ways. Then again maybe this is what Michal wants.

 

Photodynamic therapy (PDT) has been around for nearly a century and there is a large body of literature pertaining to the use of sensitizers and long wavelength visible light. That would probably be a good place to look. Of newer approaches, soft x-ray illumination using the "water window" of the nucleus might be particularly interesting, as the cells can be targeted down from whole nuclei to very tiny sub volumes (< 50 nm x 50 nm x 10 um thick). This can be done with cells at cryogenic temperatures such that the primary damage would occur via direct targeting of DNA. Non DNA free radical damage would be minimized.

 

Mario

 

 

Michal,
We have a Micropoint targeted illumination setup from Photonic Instruments for inducing DNA damage. It uses a nitrogen 365nm laser that is directly coupled to the epifluorescence path on a Nikon TE2000 inverted microscope. The laser power, number of pulses and ROI are controlled by a Metamorph plug-in.  Investigators draw a line through the nucleus with the ROI tool and lase it. They usually lase 20 to 50 cells in one session, fix and stain. Or they use the timelapse feature when their proteins are coupled to FPs.
I have no stake in the company.

Hank Adams
Genetics Department
Rm S14.8316A
U.T. M.D.Anderson Cancer Center
Houston, Tx 77030

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Michal Jarnik
Sent: Monday, June 21, 2010 11:20 AM
To: [hidden email]
Subject: Controlled DNA damage by lasers

Listers,

We would need to perform experiments including controlled DNA damage to
cells in culture. One idea was to use the LSCM and expose an area in a
way similar to the bleaching part of a FRAP experiment. We do have a LED
405 nm laser (I would assume that would be the right wavelength, we do
not have any really UV laser), Nikon C1 can do quite a lot of zooming to
concentrate the energy in a small area. I have no idea, though, about
the power necessary/available and other conditions. Any recommendations
would be very much appreciated.

Thanks, M.

--

Michal Jarnik, Ph.D.
Cell Imaging Facility
Electron Microscope Facility
Fox Chase Cancer Center
333 Cottman Ave.
Philadelphia, PA 19111
215-728-5675 (v.)
215-728-2770 (v.)
215-728-2412 (f.)

 

 

-- 

________________________________________________________________________________
Mario M. Moronne, Ph.D.

[hidden email]
[hidden email]

Gert van Cappellen Gert van Cappellen
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Controlled DNA damage by lasers

In reply to this post by Michal Jarnik
With the combination of Hoechst33342 and a 405 laser you can make controlled DNA damage. We compared this with multi-photon and 266 laser induced DNA damage.

<a href="javascript:AL_get(this, 'jour', 'J Cell Sci.');" title="Journal of cell science." style="border-bottom: 1px dotted black; color: black; font-weight: normal; text-decoration: none;">J Cell Sci. 2007 Aug 1;120(Pt 15):2731-40.

Activation of multiple DNA repair pathways by sub-nuclear damage induction methods.

Dinant C, de Jager M, Essers J, van Cappellen WA, Kanaar R, Houtsmuller AB, Vermeulen W.

Department of Pathology, Josephine Nefkens Institute, ErasmusMC, Rotterdam, The Netherlands.

Abstract

Live cell studies of DNA repair mechanisms are greatly enhanced by new developments in real-time visualization of repair factors in living cells. Combined with recent advances in local sub-nuclear DNA damage induction procedures these methods have yielded detailed information on the dynamics of damage recognition and repair. Here we analyze and discuss the various types of DNA damage induced in cells by three different local damage induction methods: pulsed 800 nm laser irradiation, Hoechst 33342 treatment combined with 405 nm laser irradiation and UV-C (266 nm) laser irradiation. A wide variety of damage was detected with the first two methods, including pyrimidine dimers and single- and double-strand breaks. However, many aspects of the cellular response to presensitization by Hoechst 33342 and subsequent 405 nm irradiation were aberrant from those to every other DNA damaging method described here or in the literature. Whereas, application of low-dose 266 nm laser irradiation induced only UV-specific DNA photo-lesions allowing the study of the UV-C-induced DNA damage response in a user-defined area in cultured cells.



on 21-06-2010 18:20 Michal Jarnik said the following:
Listers,

We would need to perform experiments including controlled DNA damage to cells in culture. One idea was to use the LSCM and expose an area in a way similar to the bleaching part of a FRAP experiment. We do have a LED 405 nm laser (I would assume that would be the right wavelength, we do not have any really UV laser), Nikon C1 can do quite a lot of zooming to concentrate the energy in a small area. I have no idea, though, about the power necessary/available and other conditions. Any recommendations would be very much appreciated.

Thanks, M.


-- 
Dr. W.A. (Gert) van Cappellen 
Room Ee914; Tel. +31-10-7043578 Mob. +31 6 22544785
Optical Imaging Centre http://www.erasmusmc.nl/oic
Reproduction & Development http://www.erasmusmc.nl/rede
Erasmus MC, 
Dr. Molenwaterplein 50
3015 GE ROTTERDAM, The Netherlands