Convenient sample for SHG

classic Classic list List threaded Threaded
♦
♦
10 messages Options
Farid Jalali Farid Jalali
Reply | Threaded
Open this post in threaded view
♦
♦
|

Convenient sample for SHG

Hello Lister,
Hope all are doing well. Can anyone please offer me suggestions for a handy sample for imaging SHG in transmission and/or reflection? Since I am not working in a lab anymore, something that I could pick up in store would be ideal. I know its a strange request but I beg your indulgence in this regard.
Cheers
Farid


Guy Cox-2 Guy Cox-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Convenient sample for SHG

There are biology shops which sell prepared slides for schools – skin would be a good choice, as would tooth or bone.  If you don’t care what you are looking at and just want to see if you are detecting a signal, anything containing starch will give a good signal.  Flour, cornstarch, laundry starch – they should all work.  If you have any access to simple chemicals, either of the sodium phosphates should be good and very bright indeed. 

 

                                                                                           Guy

 

Optical Imaging Techniques in Cell Biology

by Guy Cox    CRC Press / Taylor & Francis

     http://www.guycox.com/optical.htm

______________________________________________

Associate Professor Guy Cox, MA, DPhil(Oxon)

Australian Centre for Microscopy & Microanalysis,

Madsen Building F09, University of Sydney, NSW 2006

 

Phone +61 2 9351 3176     Fax +61 2 9351 7682

             Mobile 0413 281 861

______________________________________________

      http://www.guycox.net

 

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Farid Jalali
Sent: Saturday, 3 July 2010 6:11 AM
To: [hidden email]
Subject: Convenient sample for SHG

 

Hello Lister,
Hope all are doing well. Can anyone please offer me suggestions for a handy sample for imaging SHG in transmission and/or reflection? Since I am not working in a lab anymore, something that I could pick up in store would be ideal. I know its a strange request but I beg your indulgence in this regard.
Cheers
Farid

No virus found in this incoming message.
Checked by AVG - www.avg.com
Version: 9.0.830 / Virus Database: 271.1.1/2914 - Release Date: 07/02/10 04:35:00

George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Convenient sample for SHG

In reply to this post by Farid Jalali
Hi Farid,

A MP/SHG contact at Leica recommended urea crystals for SHG signal
for measuring the instrument response function of our FLIM component.
I don't know if you can buy this at the store (maybe online at
amazon.com?), but you could produce it yourself.

Slides containing what Guy Cox suggested can probably be purchased
from Carolina Biological Supply (www.carolina.com) or Amazon.

best wishes,
George

At 04:11 PM 7/2/2010, you wrote:
>Hello Lister,
>Hope all are doing well. Can anyone please offer me suggestions for
>a handy sample for imaging SHG in transmission and/or reflection?
>Since I am not working in a lab anymore, something that I could pick
>up in store would be ideal. I know its a strange request but I beg
>your indulgence in this regard.
>Cheers
>Farid
>
Boswell, Carl A - (cboswell) Boswell, Carl A - (cboswell)
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Convenient sample for SHG

In reply to this post by Farid Jalali
Quoting Farid Jalali <[hidden email]>:

> Hello Lister,
> Hope all are doing well. Can anyone please offer me suggestions for a handy
> sample for imaging SHG in transmission and/or reflection? Since I am not
> working in a lab anymore, something that I could pick up in store would be
> ideal. I know its a strange request but I beg your indulgence in this
> regard.
> Cheers
> Farid

Haven't tried it myself, but I've been told on good authority that a slice of
potato (lots of starch) will demonstrate SHG.  You can hardly get cheaper than
that.  I wonder if it still "performs" after it turns black from exposure to
air.

Carl Boswell
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Convenient sample for SHG

I dissolved some KDP or KTP in hot water and let it evaporate onto
slides and coverslips.  The resulting crystals give spectacular SHG
performance:

http://www.ucalgary.ca/styslab/image

Scroll down and click on 'Second Harmonic Generation Imaging' to see
my results!  The pictures were generated by imaging using a 60x or 40x
oil lens through a standard coverslip with crystals grown on the
backside using the evaporating method.  Also, there are some pictures
of rat tail collagen and dorsal column membrane with SHG signal.

Craig



On Sun, Jul 4, 2010 at 1:52 PM, Carl Boswell <[hidden email]> wrote:

> Quoting Farid Jalali <[hidden email]>:
>
>> Hello Lister,
>> Hope all are doing well. Can anyone please offer me suggestions for a
>> handy
>> sample for imaging SHG in transmission and/or reflection? Since I am not
>> working in a lab anymore, something that I could pick up in store would be
>> ideal. I know its a strange request but I beg your indulgence in this
>> regard.
>> Cheers
>> Farid
>
> Haven't tried it myself, but I've been told on good authority that a slice
> of
> potato (lots of starch) will demonstrate SHG.  You can hardly get cheaper
> than
> that.  I wonder if it still "performs" after it turns black from exposure to
> air.
>
> Carl Boswell
>
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Convenient sample for SHG

It was pointed out I didn't explain my alphabet soup.  KDP and KTP are
Potassium di- and tri-phosphate respectively.  It usually comes in a
powder.  You can image the powder directly, but the grains are small.
I prefer to dissolve the grains in hot water and grow larger crystals
to get better pictures.  Let the water evaporate on a coverslip, or
wait for a supersaturated solution to cool; large crystals will
precipitate out into solution.

Craig


On Sun, Jul 4, 2010 at 9:57 PM, Craig Brideau <[hidden email]> wrote:

> I dissolved some KDP or KTP in hot water and let it evaporate onto
> slides and coverslips.  The resulting crystals give spectacular SHG
> performance:
>
> http://www.ucalgary.ca/styslab/image
>
> Scroll down and click on 'Second Harmonic Generation Imaging' to see
> my results!  The pictures were generated by imaging using a 60x or 40x
> oil lens through a standard coverslip with crystals grown on the
> backside using the evaporating method.  Also, there are some pictures
> of rat tail collagen and dorsal column membrane with SHG signal.
>
> Craig
>
>
>
> On Sun, Jul 4, 2010 at 1:52 PM, Carl Boswell <[hidden email]> wrote:
>> Quoting Farid Jalali <[hidden email]>:
>>
>>> Hello Lister,
>>> Hope all are doing well. Can anyone please offer me suggestions for a
>>> handy
>>> sample for imaging SHG in transmission and/or reflection? Since I am not
>>> working in a lab anymore, something that I could pick up in store would be
>>> ideal. I know its a strange request but I beg your indulgence in this
>>> regard.
>>> Cheers
>>> Farid
>>
>> Haven't tried it myself, but I've been told on good authority that a slice
>> of
>> potato (lots of starch) will demonstrate SHG.  You can hardly get cheaper
>> than
>> that.  I wonder if it still "performs" after it turns black from exposure to
>> air.
>>
>> Carl Boswell
>>
>
Farid Jalali Farid Jalali
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Convenient sample for SHG

Thanks very much to everyone for the replies. As always, much appreciated.
Best
Farid

On Mon, Jul 5, 2010 at 2:48 PM, Craig Brideau <[hidden email]> wrote:
It was pointed out I didn't explain my alphabet soup.  KDP and KTP are
Potassium di- and tri-phosphate respectively.  It usually comes in a
powder.  You can image the powder directly, but the grains are small.
I prefer to dissolve the grains in hot water and grow larger crystals
to get better pictures.  Let the water evaporate on a coverslip, or
wait for a supersaturated solution to cool; large crystals will
precipitate out into solution.

Craig


On Sun, Jul 4, 2010 at 9:57 PM, Craig Brideau <[hidden email]> wrote:
> I dissolved some KDP or KTP in hot water and let it evaporate onto
> slides and coverslips.  The resulting crystals give spectacular SHG
> performance:
>
> http://www.ucalgary.ca/styslab/image
>
> Scroll down and click on 'Second Harmonic Generation Imaging' to see
> my results!  The pictures were generated by imaging using a 60x or 40x
> oil lens through a standard coverslip with crystals grown on the
> backside using the evaporating method.  Also, there are some pictures
> of rat tail collagen and dorsal column membrane with SHG signal.
>
> Craig
>
>
>
> On Sun, Jul 4, 2010 at 1:52 PM, Carl Boswell <[hidden email]> wrote:
>> Quoting Farid Jalali <[hidden email]>:
>>
>>> Hello Lister,
>>> Hope all are doing well. Can anyone please offer me suggestions for a
>>> handy
>>> sample for imaging SHG in transmission and/or reflection? Since I am not
>>> working in a lab anymore, something that I could pick up in store would be
>>> ideal. I know its a strange request but I beg your indulgence in this
>>> regard.
>>> Cheers
>>> Farid
>>
>> Haven't tried it myself, but I've been told on good authority that a slice
>> of
>> potato (lots of starch) will demonstrate SHG.  You can hardly get cheaper
>> than
>> that.  I wonder if it still "performs" after it turns black from exposure to
>> air.
>>
>> Carl Boswell
>>
>

Steffen Dietzel Steffen Dietzel
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Convenient sample for SHG

As already suggested, we did successfully use potato slices. We
actually cut them with a microtome, got some looks :-), but every
slice that is thin enough to allow forward collection should be good
for the purpose. The signal is a pain to quantify, though, since no
two starch granules are alike.

If you need something more reproducible, you might want to
disassemble a green laser pointer. I was told it contains a 1064 nm
light source which is transformed by SHG through a crystal to 532 nm.
Maybe you can isolate that crystal. According to Wikipedia it's a
crystal of
<http://en.wikipedia.org/wiki/Potassium_titanyl_phosphate>potassium
titanyl phosphate (KTP) .
http://en.wikipedia.org/wiki/Laser_pointer#Green_laser_pointer
(If you do, let me know the result, sounds interesting).

Good luck

Steffen
Peter Gabriel Pitrone Peter Gabriel Pitrone
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Convenient sample for SHG

In reply to this post by Farid Jalali
Hello Farid,

I'd like for you to try something for me... A gummy bear (or at least a very thin slice of one)! I'm sorry if this sounds crazy, but I really would like to know what happens. I can tell you that they are very birefringent, do to trial and error. Try putting one between crossed polarizers and start to squeeze its body for a good laugh, every kid I've shown it to gets a real good kick out of it...

Pete

On Jul 2, 2010, at 22:11 PM, Farid Jalali wrote:

> Hello Lister,
> Hope all are doing well. Can anyone please offer me suggestions for a handy sample for imaging SHG in transmission and/or reflection? Since I am not working in a lab anymore, something that I could pick up in store would be ideal. I know its a strange request but I beg your indulgence in this regard.
> Cheers
> Farid
>
>
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Convenient sample for SHG

Birefringent...  gummy bears!?  Will the wonders of optics never cease? @:-)
It makes sense if you think about it though; it's a gel (sort-of) so
if you put stress on its structure one way or another you will get an
order preference.

Regarding yanking crystals out of laser pointers, that's a good way to
get a cheap doubling crystal, but if you just want a test sample my
'grow crystals on a coverslip' method is a lot less work then trying
to pry the tiny dot of crystal out of one of those pointers.  KTP,
KDP, and BBO are all good doublers.  I use a commercial BBO crystal
that cost about a grand ($1k) to double our Ti:Saph now and again.  We
do fluorescence spectroscopy with it.  By tuning the Ti:Saph and
doubling it I can get ~350-520nm light.  The key thing about the BBO
crystal is it is optimal for this kind of work so I get a pretty good
conversion efficiency out of it.  A 'scrounged' crystal may not work
as well.

Craig

On Thu, Jul 8, 2010 at 9:52 AM, Peter Pitrone <[hidden email]> wrote:

> Hello Farid,
>
> I'd like for you to try something for me... A gummy bear (or at least a very thin slice of one)! I'm sorry if this sounds crazy, but I really would like to know what happens. I can tell you that they are very birefringent, do to trial and error. Try putting one between crossed polarizers and start to squeeze its body for a good laugh, every kid I've shown it to gets a real good kick out of it...
>
> Pete
>
> On Jul 2, 2010, at 22:11 PM, Farid Jalali wrote:
>
>> Hello Lister,
>> Hope all are doing well. Can anyone please offer me suggestions for a handy sample for imaging SHG in transmission and/or reflection? Since I am not working in a lab anymore, something that I could pick up in store would be ideal. I know its a strange request but I beg your indulgence in this regard.
>> Cheers
>> Farid
>>
>>
>