Kellie Beicker |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, We have a Hamamatsu ORCA Flash 4.0 (C11440-22C) that is currently mounted on to a combined microscope and AFM setup. The on-board fan creates too much noise in both fluorescence imaging and AFM force data so we would like to cool the camera with either water cooling or forced air cooling. I am curious if anyone has done this and has knowledge of the nuances of this process. For water cooling: What type of refrigerated circulation unit was used? At what temperature was the system was operated? Did you use water or antifreeze? etc. Any possible details that may not be specifically stated in the manual or that you learned from cooling your camera would be appreciated. Additionally, the AFM support recommended forced air cooling for the quietest data acquisition, but I have even less information on this. So if anyone has air cooled this type of camera or similar, I would appreciate any details. Thank you in advance for your advice, -Kellie ---------------------------------------- Kellie Beicker Physics Research Assistant Physics & Astronomy, CB 3255 The University of North Carolina at Chapel Hill Email: [hidden email] |
|
Romin, Yevgeniy/Sloan Kettering Institute |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Kellie This isn't exactly the answer to your question, but I feel you may benefit from me sharing my experiences, since we run into the problem of noise in the AFM force data as well. We also have an AFM that is coupled to a microscope. We discovered that the less things are plugged into the system while the AFM is imaging, the better results we get, since we eliminate as many potential noise sources as possible. We usually use our camera to find and focus on the are of interest, but before beginning our force measurements we turn off the cameras and unplug them, as well as turn off all the fluorescent and transmitted light sources. This is probably not optimal in your case, but that is how we dealt with this problem. It just seems to me that any sort of forced air cooling may still introduce noise in the force measurement data. Hope this helps at least somewhat Yevgeniy ________________________________________ From: Confocal Microscopy List [[hidden email]] on behalf of Kellie Beicker [[hidden email]] Sent: Thursday, June 12, 2014 12:48 PM To: [hidden email] Subject: Cooling Hamamatsu ORCA Flash 4.0 ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, We have a Hamamatsu ORCA Flash 4.0 (C11440-22C) that is currently mounted on to a combined microscope and AFM setup. The on-board fan creates too much noise in both fluorescence imaging and AFM force data so we would like to cool the camera with either water cooling or forced air cooling. I am curious if anyone has done this and has knowledge of the nuances of this process. For water cooling: What type of refrigerated circulation unit was used? At what temperature was the system was operated? Did you use water or antifreeze? etc. Any possible details that may not be specifically stated in the manual or that you learned from cooling your camera would be appreciated. Additionally, the AFM support recommended forced air cooling for the quietest data acquisition, but I have even less information on this. So if anyone has air cooled this type of camera or similar, I would appreciate any details. Thank you in advance for your advice, -Kellie ---------------------------------------- Kellie Beicker Physics Research Assistant Physics & Astronomy, CB 3255 The University of North Carolina at Chapel Hill Email: [hidden email] ===================================================================== Please note that this e-mail and any files transmitted from Memorial Sloan-Kettering Cancer Center may be privileged, confidential, and protected from disclosure under applicable law. If the reader of this message is not the intended recipient, or an employee or agent responsible for delivering this message to the intended recipient, you are hereby notified that any reading, dissemination, distribution, copying, or other use of this communication or any of its attachments is strictly prohibited. If you have received this communication in error, please notify the sender immediately by replying to this message and deleting this message, any attachments, and all copies and backups from your computer. |
|
Glenn Merrill-Skoloff |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I liquid cool mine using a Julabo F 25 circulating cooler set to 15°C. I do use an antifreeze (Julabo Thermal ED) diluted 1:10 in water but I could probably run it with just water. — Glenn Glenn Merrill-Skoloff Division of Hemostasis and Thrombosis Laboratory Manager From: Yevgeniy Romin <[hidden email]<mailto:[hidden email]>> Reply-To: Confocal Microscopy List <[hidden email]<mailto:[hidden email]>> Date: Thursday, June 12, 2014 at 1:51 PM To: "[hidden email]<mailto:[hidden email]>" <[hidden email]<mailto:[hidden email]>> Subject: Re: Cooling Hamamatsu ORCA Flash 4.0 ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Kellie This isn't exactly the answer to your question, but I feel you may benefit from me sharing my experiences, since we run into the problem of noise in the AFM force data as well. We also have an AFM that is coupled to a microscope. We discovered that the less things are plugged into the system while the AFM is imaging, the better results we get, since we eliminate as many potential noise sources as possible. We usually use our camera to find and focus on the are of interest, but before beginning our force measurements we turn off the cameras and unplug them, as well as turn off all the fluorescent and transmitted light sources. This is probably not optimal in your case, but that is how we dealt with this problem. It just seems to me that any sort of forced air cooling may still introduce noise in the force measurement data. Hope this helps at least somewhat Yevgeniy ________________________________________ From: Confocal Microscopy List [[hidden email]<mailto:[hidden email]>] on behalf of Kellie Beicker [[hidden email]<mailto:[hidden email]>] Sent: Thursday, June 12, 2014 12:48 PM To: [hidden email]<mailto:[hidden email]> Subject: Cooling Hamamatsu ORCA Flash 4.0 ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, We have a Hamamatsu ORCA Flash 4.0 (C11440-22C) that is currently mounted on to a combined microscope and AFM setup. The on-board fan creates too much noise in both fluorescence imaging and AFM force data so we would like to cool the camera with either water cooling or forced air cooling. I am curious if anyone has done this and has knowledge of the nuances of this process. For water cooling: What type of refrigerated circulation unit was used? At what temperature was the system was operated? Did you use water or antifreeze? etc. Any possible details that may not be specifically stated in the manual or that you learned from cooling your camera would be appreciated. Additionally, the AFM support recommended forced air cooling for the quietest data acquisition, but I have even less information on this. So if anyone has air cooled this type of camera or similar, I would appreciate any details. Thank you in advance for your advice, -Kellie ---------------------------------------- Kellie Beicker Physics Research Assistant Physics & Astronomy, CB 3255 The University of North Carolina at Chapel Hill Email: [hidden email]<mailto:[hidden email]> ===================================================================== Please note that this e-mail and any files transmitted from Memorial Sloan-Kettering Cancer Center may be privileged, confidential, and protected from disclosure under applicable law. If the reader of this message is not the intended recipient, or an employee or agent responsible for delivering this message to the intended recipient, you are hereby notified that any reading, dissemination, distribution, copying, or other use of this communication or any of its attachments is strictly prohibited. If you have received this communication in error, please notify the sender immediately by replying to this message and deleting this message, any attachments, and all copies and backups from your computer. ________________________________ This message is intended for the use of the person(s) to whom it may be addressed. It may contain information that is privileged, confidential, or otherwise protected from disclosure under applicable law. If you are not the intended recipient, any dissemination, distribution, copying, or use of this information is prohibited. If you have received this message in error, please permanently delete it and immediately notify the sender. Thank you. |
|
|
In reply to this post by Kellie Beicker
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hello Kellie, Here are the recommendations for cooling the ORCA-Flash4.0 camera from Hamamatsu's Technical Support team: We recommend using soft water (but not deionized water) for cooling. No need to use any anti bacterial agent, just change the cooling water every 3 months and flush out the system every 6 months. Anti-freeze is not recommended (or necessary) due to the possibility that it could corrode the internal water jacket. Set the cooling temperature of the water to +20 degrees C. The minimum flow rate should be 1 L/minute. The following two chillers have been qualified by Hamamatsu to meet the cooling requirements: Julabo HF25-ED Thermocube 400W If you have additional questions please feel free to contact us directly. Best, Mark Hobson Marketing Manager, Scientific Cameras Hamamatsu Corporation |
|
Kevin Ryan |
|
|
In reply to this post by Kellie Beicker
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Old-school and _extremely_ crude, but in past labs I've seen forced air cooling work well if the fan is mounted remotely (off the vibration-isolated table), with the airflow directed to the camera via a suspended length of dryer hose. Duct tape may (ahem) have been involved to force the airflow through the camera case. The first time I saw this was with an intensified vidicon camera, and the cooling was sufficient to keep the black-level and gain constant on a strictly analog device not intended for microscopy. The dryer hose isolates the microscope from any fan vibrations, and the setup can provide consistent if limited cooling. Kevin Ryan Senior Project Manager Media Cybernetics, Inc. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Kellie Beicker Sent: Thursday, June 12, 2014 12:48 PM To: [hidden email] Subject: Cooling Hamamatsu ORCA Flash 4.0 ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, We have a Hamamatsu ORCA Flash 4.0 (C11440-22C) that is currently mounted on to a combined microscope and AFM setup. The on-board fan creates too much noise in both fluorescence imaging and AFM force data so we would like to cool the camera with either water cooling or forced air cooling. I am curious if anyone has done this and has knowledge of the nuances of this process. For water cooling: What type of refrigerated circulation unit was used? At what temperature was the system was operated? Did you use water or antifreeze? etc. Any possible details that may not be specifically stated in the manual or that you learned from cooling your camera would be appreciated. Additionally, the AFM support recommended forced air cooling for the quietest data acquisition, but I have even less information on this. So if anyone has air cooled this type of camera or similar, I would appreciate any details. Thank you in advance for your advice, -Kellie ---------------------------------------- Kellie Beicker Physics Research Assistant Physics & Astronomy, CB 3255 The University of North Carolina at Chapel Hill Email: [hidden email] |