Seb Stephens |
|
|
We have the problem of our cells differentiating well on specific coverslips,
and not those of mattek dishes. In the meantime of trying to identify other dishes, we continue to use coverslips. We image just with a dipping lens, but would like some more resolution. Is it possible to mount the "coverslip with cells" upsidedown onto a new mattek dish and then image it using an inverted microscope with a water immersion lens? Is there another way around this? Regards |
|
Roman Veith |
|
|
Hi Waldo,
we use custom made coverslip holder built in our toolshop. This works very well for us. Cells grow on coverslips and whenever needed single coverslips with cells are collected from the cell culture. The coverslip holder is simple a metal device with an adapter for coverslips in the middle. Imagine the metal device as a metal object slide with a hole in the middle, designed in a way, that the coverslip exactly fits into it, without passing through. E.g. for a 18x18 coverslip the hole has dimensions of ~16x16. Another piece of metal with a ~16 circular whole in it and a rubber seal is placed on top. Immidiately after transfer of the coverslips new buffer is added. The rubber seal of the upper device assures that everything is leak-proof. Maybe after this procedure some recovery time for the cells is appropriate. Cells can then be imaged even with an oil immersion objective - if refractive index mismatch is no problem... Regards, Roman Waldo Schmidt schrieb: > We have the problem of our cells differentiating well on specific coverslips, > and not those of mattek dishes. In the meantime of trying to identify other > dishes, we continue to use coverslips. > > We image just with a dipping lens, but would like some more resolution. > > Is it possible to mount the "coverslip with cells" upsidedown onto a new > mattek dish and then image it using an inverted microscope with a water > immersion lens? > > Is there another way around this? > > Regards > -- -- Roman Veith Dipl.-Biol. Institut für physikalische und theoretische Chemie AG Biophysikalische Chemie Universität Bonn Wegelerstraße 12 53115 Bonn Tel-Nr.: 0228-733089 Fax-Nr.: 0228-739424 http://www.thch.uni-bonn.de/pc/kubitscheck/ |
|
Pascal Lorentz-2 |
|
|
Hi Waldo,
I think what Roman describes is also sold by a company in Switzerland. They are called Life Imaging Services: http://www.lis.ch/ No commercial interest, just a satisfied customer. Best regards Pascal Roman Veith schrieb: > Hi Waldo, > > we use custom made coverslip holder built in our toolshop. This works > very well for us. Cells grow on coverslips and whenever needed single > coverslips with cells are collected from the cell culture. The > coverslip holder is simple a metal device with an adapter for > coverslips in the middle. Imagine the metal device as a metal object > slide with a hole in the middle, designed in a way, that the coverslip > exactly fits into it, without passing through. E.g. for a 18x18 > coverslip the hole has dimensions of ~16x16. Another piece of metal > with a ~16 circular whole in it and a rubber seal is placed on top. > Immidiately after transfer of the coverslips new buffer is added. The > rubber seal of the upper device assures that everything is leak-proof. > Maybe after this procedure some recovery time for the cells is > appropriate. > Cells can then be imaged even with an oil immersion objective - if > refractive index mismatch is no problem... > > Regards, > Roman > > > Waldo Schmidt schrieb: >> We have the problem of our cells differentiating well on specific >> coverslips, and not those of mattek dishes. In the meantime of trying >> to identify other dishes, we continue to use coverslips. >> >> We image just with a dipping lens, but would like some more resolution. >> >> Is it possible to mount the "coverslip with cells" upsidedown onto a >> new mattek dish and then image it using an inverted microscope with a >> water immersion lens? >> Is there another way around this? >> Regards >> > > Pascal Lorentz Department of Biomedicine University of Basel Mattenstrasse 28 4058 Basel Switzerland |
|
Cameron Nowell |
|
|
In reply to this post by Seb Stephens
Hi Waldo,
Your other choice is to get a Sykes-Moore chanmber made by Belco. The are a simple chamber with a rubber seal and a screw on top that holds it all together. We use them daily and they work great. You can grow your cells on the coverslips then just mount them in the chamber and image happily. The only problem i have had with them is that you cannot usually image to whole coverslip. When mounted the cover slip is recessed a little bit which means large objectives can only focus on the middle area of the cover slip. A big thanks to Steve Cody for buying these in for the Centre. Also I have no comercial intrsts with Belco. Cheers Cam Cameron J. Nowell Microscpy Manager Central Resource for Advanced Microscopy Ludwig Insttue for Cancer Research PO Box 2008 Royal Melbourne Hospital Victoria, 3050 AUSTRALIA Office: +61 3 9341 3155 Mobile: +61422882700 Fax: +61 3 9341 3104 http://www.ludwig.edu.au/branch/research/platform/microscopy.htm ________________________________ From: Confocal Microscopy List on behalf of Waldo Schmidt Sent: Wed 26/08/2009 11:31 PM To: [hidden email] Subject: Coverslip on coverslip - can cells inbetween be imaged? We have the problem of our cells differentiating well on specific coverslips, and not those of mattek dishes. In the meantime of trying to identify other dishes, we continue to use coverslips. We image just with a dipping lens, but would like some more resolution. Is it possible to mount the "coverslip with cells" upsidedown onto a new mattek dish and then image it using an inverted microscope with a water immersion lens? Is there another way around this? Regards |
|
Jacqueline Ross |
|
|
In reply to this post by Seb Stephens
Hi Waldo,
I know others have already replied with some suggestions but I have one more for you to consider. I recently bought two chambers from Quorum Technologies, called Chamlide chambers. They are designed to take various sizes of coverslip ( I bought the 25mm one) and are magnetic. I have only just bought them having seen them at the Vancouver LiveCell course and haven't yet tried them but they seem like they will work really well. No commercial interest! Kind regards, Jacqui Jacqueline Ross Biomedical Imaging Microscopist Biomedical Imaging Research Unit School of Medical Sciences Faculty of Medical & Health Sciences The University of Auckland Private Bag 92019 Auckland, NEW ZEALAND Tel: 64 9 373 7599 Ext 87438 Fax: 64 9 373 7484 http://www.fmhs.auckland.ac.nz/sms/biru/ -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Waldo Schmidt Sent: Thursday, 27 August 2009 1:31 a.m. To: [hidden email] Subject: Coverslip on coverslip - can cells inbetween be imaged? We have the problem of our cells differentiating well on specific coverslips, and not those of mattek dishes. In the meantime of trying to identify other dishes, we continue to use coverslips. We image just with a dipping lens, but would like some more resolution. Is it possible to mount the "coverslip with cells" upsidedown onto a new mattek dish and then image it using an inverted microscope with a water immersion lens? Is there another way around this? Regards |
|
Alberto Diaspro |
|
|
Dear friends,
in my department at the Italian Institute of Technology there are 20 PhD positions open. A specific one is reported here. The complete call can be found at www.iit.it. All the best Alby ----- Diaspro Lab call ---- Theme 3.1: Optical Nanoscopy and advanced high-resolution microscopy. Tutor: Alberto Diaspro A recognized advantage of optical microscopy lies in the fact that allows non-invasive three-dimensional (3D) imaging of live cells at the submicron scale with high specificity. The advent of the visible fluorescent proteins and of a myriad of fluorescent tags pushed fluorescence microscopy to become the most popular imaging tool in cell biology. The confocal and multiphoton versions of fluorescence microscopy reinforce this condition. However, like any other standard imaging system relying on focused light, all these microscopes are limited in spatial resolution because the smallest possible spot size is imposed by diffraction. Several approaches aimed at overcoming the diffraction limit. Stimulated Emission Depletion (STED) microscopy is one of the most promising Optical Nanoscopy approaches allowing an ultimate resolution of 7.6 nm in the optical regime. In STED microscopy, fluorescence emanating from the periphery of the focused excitation beam is suppressed by a second properly shaped beam that depletes the excited state population through stimulated emission. This effectively narrows fluorescent molecule signature, the point spread function (PSF) of the microscope, to permit superresolved images to be acquired. The Optical Nanoscopy theme is related to the development of an original STED architecture based in white light laser illumination and to its characterization for key applications to tracking of molecular events in living cells towards the study of degenerative processes like neuro-diseases, tumor progression and nanoparticle toxic effects on biological systems. Variations on the theme will include Fluorescence Photo- Activation Light Microscopy (FPALM) and Single Plane Illumination Microscopy (SPIM) approaches. For further details concerning the research project, please contact: [hidden email] ------ ---------------------------------------- Alberto Diaspro Head, Nanophysics Unit Senior Scientist The Italian Institute of Technology -IIT Via Morego, 30 16163 - Genova (Italy) phone: +39 010 71781503 mobile: +393666719968 fax: +39 010 720321 http://www.iit.it [hidden email] Professor of Applied Physics Department of Physics University of Genova Via Dodecaneso, 33 16146 Genova - Italy tel. +39 010 353 6426 fax. +39 010 314218 http://www.lambs.it [hidden email] ------------------------------------------------------- |
|
Seb Stephens |
|
|
In reply to this post by Jacqueline Ross
Hi Jacqui,
waldo = me!!!!!!!! I have just ordered these chamlide chambers too. You know theyre on special now? Like 50% off.... Speak soon Sébastien Stephens
Harvard University School of Dental Medicine 188 Longwood Ave REB 313 Boston, MA 02115 Ph: 617-4327328 Fax: 617-4321897 [hidden email] > Date: Thu, 27 Aug 2009 09:49:57 +1200 > From: [hidden email] > Subject: Re: Coverslip on coverslip - can cells inbetween be imaged? > To: [hidden email] > > Hi Waldo, > > I know others have already replied with some suggestions but I have one > more for you to consider. I recently bought two chambers from Quorum > Technologies, called Chamlide chambers. They are designed to take > various sizes of coverslip ( I bought the 25mm one) and are magnetic. I > have only just bought them having seen them at the Vancouver LiveCell > course and haven't yet tried them but they seem like they will work > really well. > > No commercial interest! > > Kind regards, > > Jacqui > > Jacqueline Ross > > Biomedical Imaging Microscopist > Biomedical Imaging Research Unit > School of Medical Sciences > Faculty of Medical & Health Sciences > The University of Auckland > Private Bag 92019 > Auckland, NEW ZEALAND > > Tel: 64 9 373 7599 Ext 87438 > Fax: 64 9 373 7484 > > http://www.fmhs.auckland.ac.nz/sms/biru/ > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Waldo Schmidt > Sent: Thursday, 27 August 2009 1:31 a.m. > To: [hidden email] > Subject: Coverslip on coverslip - can cells inbetween be imaged? > > We have the problem of our cells differentiating well on specific > coverslips, > and not those of mattek dishes. In the meantime of trying to identify > other > dishes, we continue to use coverslips. > > We image just with a dipping lens, but would like some more resolution. > > Is it possible to mount the "coverslip with cells" upsidedown onto a new > > mattek dish and then image it using an inverted microscope with a water > immersion lens? > > Is there another way around this? > > Regards Partagez vos souvenirs sur le Web avec les personnes de votre choix les personnes de votre choix. |
|
|
I tried today mounting a "coverslip+cells" upsidedown on another coverslip, all covered in media...
Cells died in a few hours.... Partagez vos souvenirs sur le Web avec les personnes de votre choix les personnes de votre choix. |
|
Mayandi Sivaguru |
|
|
Try, ibidi dishes ibidi.com no commercial interest.
Shiv > I tried today mounting a "coverslip+cells" upsidedown on another > coverslip, all covered in media... > > > > Cells died in a few hours.... > > _________________________________________________________________ > Partagez vos souvenirs sur le Web avec les personnes de votre choix. > http://www.microsoft.com/northafrica/windows/windowslive/products/photos-share.aspx?tab=1 |
|
Keith Morris |
|
|
I'd also recommend the trying the plastic cover-slip thickness membrane
slides and wells from Ibidi: Ibidi μ-well MultiWell, micro-channel and Flask 3x1" slides [new product] for slide cell culture and hi-res microscopy imaging. Go to Ibidis web site. Wells and channels with a 150μm membrane cover-slip on a standard slide http://www.ibidi.de/products/p_disposables.html Griener also make utra-thin Lumox membrane [as well as glass or plastic] wells and slides for inverted microscope viewing: Greiner bio-one MultiWell and SlideFlask slides [new product] for slide cell culture and microscopy imaging. Download Greiner Culture Slide Brochure. 1 to 8 wells on a Lumox membrane, cover-slip or glass slide http://www.well.ox.ac.uk/cytogenetics/downloads/Griener MultiWell & Flask Slides.pdf You can of course buy the Mattek dishes as uncoated, poly-d-lysine coated, or collagen coated, plus they offer two cover-slip surfaces [standard glass and plastic CS1 - the latter's unsuitable for oil immersion though] that might help. http://www.glass-bottom-dishes.com/faq.html#typedish http://www.glass-bottom-dishes.com/coverslips.html Our workshop back at the Institute of Ophthalmology, UCL, did try making our own 'Mattek' style dishes with coverslips of our choice, but it was hard cutting out the circle at the base of the Petri dish and they gave up. I did consider laser cutting out-sourcing to make the hole [that we used very successfully for cutting our plastic CR-39 autoradiography 3x1" slides years ago] and applying the cover-slip of choice, but it was easier just buying in the Mattek dishes as they generally worked OK for our cell biologists. Most suppliers offer free samples to try out. Regards Keith For more web-links see http://www.well.ox.ac.uk/external-website-links --------------------------------------------------------------------------- Dr Keith J. Morris, Molecular Cytogenetics and Microscopy Core, Laboratory 00/069 and 00/070, The Wellcome Trust Centre for Human Genetics, Roosevelt Drive, Oxford OX3 7BN, United Kingdom. Telephone: +44 (0)1865 287568 Email: [hidden email] Web-pages: http://www.well.ox.ac.uk/molecular-cytogenetics-and-microscopy -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email] Sent: 27 August 2009 03:59 To: [hidden email] Subject: Re: Coverslip on coverslip - can cells inbetween be imaged? Try, ibidi dishes ibidi.com no commercial interest. Shiv > I tried today mounting a "coverslip+cells" upsidedown on another > coverslip, all covered in media... > > > > Cells died in a few hours.... > > _________________________________________________________________ > Partagez vos souvenirs sur le Web avec les personnes de votre choix. > re.aspx?tab=1 |
|
|
In reply to this post by Seb Stephens
maybe use a couple of coverslip as spacer, so the cells won't be under pressure.
good luck.
Lai
From: Confocal Microscopy List [[hidden email]] On Behalf Of Seb Stephens [[hidden email]] Sent: Wednesday, August 26, 2009 9:56 PM To: [hidden email] Subject: I tried today mounting a "coverslip+cells" upsidedown on another coverslip, all covered in media...
Cells died in a few hours.... Partagez vos souvenirs sur le Web avec les personnes de votre choix les personnes de votre choix. |
|
Ignatius, Mike-2 |
|
|
In reply to this post by Cameron Nowell
Hi Waldo,
One more suggestion. We have been selling the AttoFluor chamber (A7816) for over a decade for this application as well. Your coverslip (25mm round) and our holder. Very simple, autoclavable design. Regards, Mike Ignatius, Molecular Probes/Life Technologies. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Cameron Nowell Sent: Wednesday, August 26, 2009 2:18 PM To: [hidden email] Subject: Re: Coverslip on coverslip - can cells inbetween be imaged? Hi Waldo, Your other choice is to get a Sykes-Moore chanmber made by Belco. The are a simple chamber with a rubber seal and a screw on top that holds it all together. We use them daily and they work great. You can grow your cells on the coverslips then just mount them in the chamber and image happily. The only problem i have had with them is that you cannot usually image to whole coverslip. When mounted the cover slip is recessed a little bit which means large objectives can only focus on the middle area of the cover slip. A big thanks to Steve Cody for buying these in for the Centre. Also I have no comercial intrsts with Belco. Cheers Cam Cameron J. Nowell Microscpy Manager Central Resource for Advanced Microscopy Ludwig Insttue for Cancer Research PO Box 2008 Royal Melbourne Hospital Victoria, 3050 AUSTRALIA Office: +61 3 9341 3155 Mobile: +61422882700 Fax: +61 3 9341 3104 http://www.ludwig.edu.au/branch/research/platform/microscopy.htm ________________________________ From: Confocal Microscopy List on behalf of Waldo Schmidt Sent: Wed 26/08/2009 11:31 PM To: [hidden email] Subject: Coverslip on coverslip - can cells inbetween be imaged? We have the problem of our cells differentiating well on specific coverslips, and not those of mattek dishes. In the meantime of trying to identify other dishes, we continue to use coverslips. We image just with a dipping lens, but would like some more resolution. Is it possible to mount the "coverslip with cells" upsidedown onto a new mattek dish and then image it using an inverted microscope with a water immersion lens? Is there another way around this? Regards |