Teemu Ihalainen |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear list, I have a small question considering coverslip thickness correction when using water immersion objectives and I was hoping that you could help me out. We have bought a water immersion objective from Olympus (UPLSAPO 60x, 1.2 NA) with a correction collar. The objective works nicely in our LSM510, when imaging living cells it gives good images, the PSF is better compared to oil immersion objectives, etc. So, everything goes according to the theory. There is only one problem. The objecive is very "picky" about the coverslip thickness (again, like the theory predicts...). We are using glass bottom dishes from MatTek which have the thickness in range of 160-190um. Is there some simple way to adjust the correction collar when imaging your sample? The adjustment is easy to do when you are imaging for example ps-specks but when you have your real sample, it is trickier. I have thought/found a couple of ways to do it (below) but Im not sure which is the best and are there other ways to do it. I Measure of the glass thickness before you culture your cells. Then at the microscope you just adjust the collar accordingly. II At the microscope you first try to find a really small detail(s) in your sample and then adjust the collar for best image quality (image z-stacks or something like that). III Image some cell adhering beads(?) what you can just add to your sample. Unfortunately I dont know if such beads even exists. IV Buy some other dishes with more accurate glass thickness. Is there any? Any suggestions and tips are warmly welcome. Best Regards, Teemu Ihalainen ------------------------------------------------- M.Sc., Nanoscience University of Jyväskylä Department of Biological and Environmental science Molecular Biology, room B 212.2 Survontie 9 B2, 40500 Jyväskylä, Finland Tel. +358-14-260 4158 Mobile +358-50-518 7422 ------------------------------------------------- |
|
Michael Weber-4 |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hey Teemu, I tried several ways of adjusting the correction collar, and for me the results are as follows: - The numbers written at the collar are not very precise. However, measuring the thickness of coverslip and then put it to the according number brings you at least in the range of the correct value. - The way which brings me the best results is to perform a line Z-stack over the coverslip, and go for the thinnest representation of it at half maximum. Big disadvantage: this technique is quite a pain without fast Z-stage. - An alternative way is to simply go for the highest intensity while imaging a sample in XY and using a line profile. To my understanding, when it comes to refractive index missmatches, you lose light due to refraction. If it matches better, you get more light back into the objective, thus the best correction collar position should give you the highest intensity. I would appreciate if somebody could comment on that. cheers, Michael Teemu Ihalainen wrote: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Dear list, > > I have a small question considering coverslip thickness correction when > using water immersion objectives and I was hoping that you could help me > out. > > We have bought a water immersion objective from Olympus (UPLSAPO 60x, 1.2 > NA) with a correction collar. The objective works nicely in our LSM510, > when imaging living cells it gives good images, the PSF is better compared > to oil immersion objectives, etc. So, everything goes according to the > theory. > > There is only one problem. The objecive is very "picky" about the > coverslip thickness (again, like the theory predicts...). We are using > glass bottom dishes from MatTek which have the thickness in range of > 160-190um. Is there some simple way to adjust the correction collar when > imaging your sample? > > The adjustment is easy to do when you are imaging for example ps-specks > but when you have your real sample, it is trickier. I have thought/found a > couple of ways to do it (below) but Im not sure which is the best and are > there other ways to do it. > > I > Measure of the glass thickness before you culture your cells. Then at the > microscope you just adjust the collar accordingly. > > II > At the microscope you first try to find a really small detail(s) in your > sample and then adjust the collar for best image quality (image z-stacks > or something like that). > > III > Image some cell adhering beads(?) what you can just add to your sample. > Unfortunately I dont know if such beads even exists. > > IV > Buy some other dishes with more accurate glass thickness. Is there any? > > Any suggestions and tips are warmly welcome. > > > > Best Regards, > Teemu Ihalainen > ------------------------------------------------- > M.Sc., Nanoscience > University of Jyväskylä > Department of Biological and Environmental science > Molecular Biology, room B 212.2 > Survontie 9 B2, 40500 Jyväskylä, Finland > Tel. +358-14-260 4158 > Mobile +358-50-518 7422 > ------------------------------------------------- |
|
Armstrong, Brian |
|
|
In reply to this post by Teemu Ihalainen
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Well, I think you should start with method I, and then use method II for fine tuning. After all it is the image quality you are after regardless of what numbers on the collar are used. It is my understanding that Zeiss uses the tube lens for correction whereas Olympus corrects entirely in the objective (see Handbook of Biological Confocal Microscopy, Pawley J). Which begs the question, why are you putting an Olympus lens on your LSM510? Try the Zeiss 63x/1.2W Corr and see if you have the same issues. Cheers, Brian D Armstrong PhD Light Microscopy Core Manager Beckman Research Institute City of Hope 1450 E Duarte Rd Duarte, CA 91010 626-359-8111 x62872 http://www.cityofhope.org/SharedResources/LightMicroscopy -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Teemu Ihalainen Sent: Thursday, December 06, 2007 5:26 AM To: [hidden email] Subject: Coverslip thickness and correction collar Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear list, I have a small question considering coverslip thickness correction when using water immersion objectives and I was hoping that you could help me out. We have bought a water immersion objective from Olympus (UPLSAPO 60x, 1.2 NA) with a correction collar. The objective works nicely in our LSM510, when imaging living cells it gives good images, the PSF is better compared to oil immersion objectives, etc. So, everything goes according to the theory. There is only one problem. The objecive is very "picky" about the coverslip thickness (again, like the theory predicts...). We are using glass bottom dishes from MatTek which have the thickness in range of 160-190um. Is there some simple way to adjust the correction collar when imaging your sample? The adjustment is easy to do when you are imaging for example ps-specks but when you have your real sample, it is trickier. I have thought/found a couple of ways to do it (below) but Im not sure which is the best and are there other ways to do it. I Measure of the glass thickness before you culture your cells. Then at the microscope you just adjust the collar accordingly. II At the microscope you first try to find a really small detail(s) in your sample and then adjust the collar for best image quality (image z-stacks or something like that). III Image some cell adhering beads(?) what you can just add to your sample. Unfortunately I dont know if such beads even exists. IV Buy some other dishes with more accurate glass thickness. Is there any? Any suggestions and tips are warmly welcome. Best Regards, Teemu Ihalainen ------------------------------------------------- M.Sc., Nanoscience University of Jyväskylä Department of Biological and Environmental science Molecular Biology, room B 212.2 Survontie 9 B2, 40500 Jyväskylä, Finland Tel. +358-14-260 4158 Mobile +358-50-518 7422 ------------------------------------------------- --------------------------------------------------------------------- SECURITY/CONFIDENTIALITY WARNING: This message and any attachments are intended solely for he individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. --------------------------------------------------------------------- |
|
|
In reply to this post by Teemu Ihalainen
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal If it wasn't 'picky' it wouldn't be doing its job! Correct adjustment is crucial and without it you'll be doing worse than using a lower NA lens. Forget looking for more accurate cover slips - at this level sufficiently accurate ones don't exist so all it would do is give you a closer starting point. Method 2 is the way to go but the key is not to look for best image quality (even though this is what you'll obtain, it's hard to spot). What you are looking for is IDENTICAL IMAGES EITHER SIDE OF FOCUS. If your spot goes to a circle one side of focus and a spot with a halo around it on the other side, you have SA. Adjust the collar (widefield, visual observation) till the image looks identical either side of focus. It will actually go to a bit of a ring on either side, and this is normal (a look at a diagram of the psf will show you why). It's easiest to do this in fluorescence but if fading is a problem it can also be done in brightfield. Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________ Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Teemu Ihalainen Sent: Friday, 7 December 2007 12:26 AM To: [hidden email] Subject: Coverslip thickness and correction collar Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear list, I have a small question considering coverslip thickness correction when using water immersion objectives and I was hoping that you could help me out. We have bought a water immersion objective from Olympus (UPLSAPO 60x, 1.2 NA) with a correction collar. The objective works nicely in our LSM510, when imaging living cells it gives good images, the PSF is better compared to oil immersion objectives, etc. So, everything goes according to the theory. There is only one problem. The objecive is very "picky" about the coverslip thickness (again, like the theory predicts...). We are using glass bottom dishes from MatTek which have the thickness in range of 160-190um. Is there some simple way to adjust the correction collar when imaging your sample? The adjustment is easy to do when you are imaging for example ps-specks but when you have your real sample, it is trickier. I have thought/found a couple of ways to do it (below) but Im not sure which is the best and are there other ways to do it. I Measure of the glass thickness before you culture your cells. Then at the microscope you just adjust the collar accordingly. II At the microscope you first try to find a really small detail(s) in your sample and then adjust the collar for best image quality (image z-stacks or something like that). III Image some cell adhering beads(?) what you can just add to your sample. Unfortunately I dont know if such beads even exists. IV Buy some other dishes with more accurate glass thickness. Is there any? Any suggestions and tips are warmly welcome. Best Regards, Teemu Ihalainen ------------------------------------------------- M.Sc., Nanoscience University of Jyväskylä Department of Biological and Environmental science Molecular Biology, room B 212.2 Survontie 9 B2, 40500 Jyväskylä, Finland Tel. +358-14-260 4158 Mobile +358-50-518 7422 ------------------------------------------------- No virus found in this incoming message. Checked by AVG Free Edition. Version: 7.5.503 / Virus Database: 269.16.17/1177 - Release Date: 7/12/2007 1:11 PM No virus found in this outgoing message. Checked by AVG Free Edition. Version: 7.5.503 / Virus Database: 269.16.17/1177 - Release Date: 7/12/2007 1:11 PM |
|
George McNamara |
|
|
In reply to this post by Teemu Ihalainen
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
option Ib. Measure of the glass thickness before you culture your cells. Then, for this lens, use dishes whose coverglasses are 170 um (or whatever thickness works best for the lens at its current "optimal" setting). If one in five are the right thickness, than the cost of that dish, for this experiment, has gone from about $2 to $10. Should take less than a minute to measure a dish using a 20x, decent NA, dry lens. Compared to the cost of the confocal time (or of an LSM510), or of your time to do manual image analysis, or cost of the analysis computer and image analysis software, the extra cost is trivial. Plus, you can sort the "rejects" and use them with lenses each thickness is optimized for. At 08:25 AM 12/6/2007, you wrote: Search the CONFOCAL archive at George McNamara, Ph.D. University of Miami, Miller School of Medicine Image Core Miami, FL 33010 [hidden email] [hidden email] 305-243-8436 office http://home.earthlink.net/~pubspectra/ http://home.earthlink.net/~geomcnamara/ http://www.sylvester.org/health_pro/shared_resources/index.asp (see Analytical Imaging Core Facility) |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
But ... but .....
(a) this will be nowhere near accurate
enough
(b) setting the ring manually once you know what to
look for takes only a minute or two
If you are paying more than $8 per minute for
confocal
time you are being seriously
overcharged!
Guy
Optical Imaging Techniques in Cell Biology From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of George McNamara Sent: Sunday, 9 December 2007 6:08 AM To: [hidden email] Subject: Re: Coverslip thickness and correction collar ... option Ib option Ib. Measure of the glass thickness before you culture your cells. Then, for this lens, use dishes whose coverglasses are 170 um (or whatever thickness works best for the lens at its current "optimal" setting). If one in five are the right thickness, than the cost of that dish, for this experiment, has gone from about $2 to $10. Should take less than a minute to measure a dish using a 20x, decent NA, dry lens. Compared to the cost of the confocal time (or of an LSM510), or of your time to do manual image analysis, or cost of the analysis computer and image analysis software, the extra cost is trivial. Plus, you can sort the "rejects" and use them with lenses each thickness is optimized for. At 08:25 AM 12/6/2007, you wrote: Search the CONFOCAL archive at George McNamara, Ph.D. University of Miami, Miller School of Medicine Image Core Miami, FL 33010 [hidden email] [hidden email] 305-243-8436 office http://home.earthlink.net/~pubspectra/ http://home.earthlink.net/~geomcnamara/ http://www.sylvester.org/health_pro/shared_resources/index.asp (see Analytical Imaging Core Facility) No virus found in this incoming message. No virus found in this outgoing message. |
|
diaspro1959@katamail.com |
|
|
In reply to this post by George McNamara
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal INTERNATIONAL SCHOOL OF BIOPHYSICS «ANTONIO BORSELLINO» 36th Course: MULTIDIMENSIONAL OPTICAL FLUORESCENCE MICROSCOPY TOWARDS NANOSCOPY ERICESICILY: 19 29 APRIL 2008 Sponsored by the: Italian Ministry of Education, University and Scientific Research Sicilian Regional Government, Companies. LECTURERS Fluorescence Spectroscopy, GFP Photophysics R. BIZZARRI, NEST-INFM, SNS, Pisa, IT; Optics, Confocal Microscopy, THG F. BRAKENHOFF, University of Amsterdam, NL; FRAP, Single particle tracking K. BRAECKMANS, Ghent University, BE; Single molecule force spectroscopy J. BRUJIC, New York University, USA; Fluctuation Microscopies for biological tissues GIUSEPPE CHIRICO, University of Milan-Bicocca, IT; Micro-particle manipulation D. COJOC, TASC, INFM, Trieste, IT; SHG, CARS, 2PE C. COMBS, NIH, Bethesda, USA; 2PE, 3D imaging A. DIASPRO, University of Genoa, IT; Micro/Nano Optical Manipulation E. Di FABRIZIO, Unversity of Catanzaro Magna Graecia, IT; Raster Image Correlation Spectroscopy, Photon Counting M. DIGMAN, UC Irvine, USA; High-content screening M. FARETTA, IFOM-IEO, Milan, IT; Correlative Microscopy U. FASCIO, University of Milan, IT; Fluorescence Lifetime, FRET H.C. GERRITSEN, Utrecht University, NL; FCS, Global Data Analysis E. GRATTON, UC Irvine, USA; Photonic crystals, nanophoton ics M. GU, Swinburne Univ. of Technology, Victoria, AU; Time lapse imaging S. GUIDO, Universit of Naples, IT; Fluorescence Optical Nanoscopy S. HELL, MPI, Goettingen, DE; Scanning Microscopy, Optical aberrations M. MARTINEZ CORRAL, Univ. of Valencia, ES; Optical systems, Scanning Microscopy F. QUERCIOLI, CNR-ISC, Florence, IT; 2PE, Fast scanning methods P. SAGGAU, Baylor College of Med. Hou ston, Texas, USA; Correlative Microscopy at cryo-Temperatures A. SARTORI, Institut Pasteur, Paris, FR; Light Scattering, FCS applications P.L. SAN BIAGIO, CNR-IBF, Palermo, IT; Molecular landscapes by means of AFM G. SCOLES, Princeton University, USA; Linear and Non linear Optical Microscopy C. SHEPPARD, Ntl Univ. of Singapore, Singapore; Optical Microscopy, 3D imaging, Photonic Forces E. STELZER, EMBL, Heidelberg, DE; Laser scissors and tweezers in cell biology I. TOLIC-NORRELYKKE, MPI, Dresden, DE; Fluorescence imaging in Neuroscience V. TORRE, SISSA, Trieste, IT; Quantitative colocalization C. USAI, CNR-IBF, Genoa, IT; Confocal Microscopy, Structured light methods T. WILSON, University of Oxford, UK; Photoswitch-activatable fluorescent proteins, Lifetime F. WOUTERS, Univ. of Goettingen, DE. -------- ---------------------------------------------------- "Follow knowledge wherever it leads us." (reading Galileo Galilei) ----------------------------------------------------- Alberto Diaspro, EBSA President-Elect, MicroScoBIO LAMBS-IFOM, Department of Physics, University of Genoa, Via Dodecaneso 33, 16146 Genoa, Italy - fax +39-010314218 - tel +39 0103536426/309; URLs: www.lambs.it; EBSA is Biophysics in Europe - European Biophysical Societies' Association www.ebsa.org ---------------------------------------------- ________________________________________________________________________ SERVIZIO VOICE: TELEFONA e INVIA SMS dal tuo computer a tariffe vantaggiose! Scopri come telefonare e videochiamare gratis da pc a pc. http://voice.repubblica.it |
|
anna boccaccio |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal ----- Original Message ----- From: "[hidden email]" <[hidden email]> To: <[hidden email]> Sent: Sunday, December 09, 2007 1:38 PM Subject: Alby - after FOM 2008 in Japan --- Erice 2008, Sicily > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > INTERNATIONAL SCHOOL OF BIOPHYSICS «ANTONIO BORSELLINO» 36th Course: > MULTIDIMENSIONAL OPTICAL FLUORESCENCE MICROSCOPY TOWARDS NANOSCOPY > ERICE-SICILY: 19 - 29 APRIL 2008 > Sponsored by the: . Italian Ministry of Education, University and > Scientific Research . Sicilian Regional Government, Companies. > > LECTURERS > Fluorescence Spectroscopy, GFP Photophysics > . R. BIZZARRI, NEST-INFM, SNS, Pisa, IT; Optics, Confocal Microscopy, THG > . F. BRAKENHOFF, University of Amsterdam, NL; FRAP, Single particle > tracking . K. BRAECKMANS, Ghent University, BE; Single molecule force > spectroscopy . J. BRUJIC, New York University, USA; Fluctuation > Microscopies for biological tissues GIUSEPPE CHIRICO, University of > Milan-Bicocca, IT; Micro-particle manipulation . D. COJOC, TASC, INFM, > Trieste, IT; SHG, CARS, 2PE . C. COMBS, NIH, Bethesda, USA; 2PE, 3D > imaging . A. DIASPRO, University of Genoa, IT; Micro/Nano Optical > Manipulation . E. Di FABRIZIO, Unversity of Catanzaro Magna Graecia, IT; > Raster Image Correlation Spectroscopy, Photon Counting . M. DIGMAN, UC > Irvine, USA; High-content screening . M. FARETTA, IFOM-IEO, Milan, IT; > Correlative Microscopy . U. FASCIO, University of Milan, IT; Fluorescence > Lifetime, FRET . H.C. GERRITSEN, Utrecht University, NL; FCS, Global Data > Analysis . E. GRATTON, UC Irvine, USA; Photonic crystals, nanophoton > > ics . M. GU, Swinburne Univ. of Technology, Victoria, AU; Time lapse > imaging . S. GUIDO, Universit of Naples, IT; Fluorescence Optical > Nanoscopy . S. HELL, MPI, Goettingen, DE; Scanning Microscopy, Optical > aberrations . M. MARTINEZ CORRAL, Univ. of Valencia, ES; Optical systems, > Scanning Microscopy . F. QUERCIOLI, CNR-ISC, Florence, IT; 2PE, Fast > scanning methods . P. SAGGAU, Baylor College of Med. Hou > ston, Texas, USA; Correlative Microscopy at cryo-Temperatures . A. > SARTORI, Institut Pasteur, Paris, FR; Light Scattering, FCS applications > . P.L. SAN BIAGIO, CNR-IBF, Palermo, IT; Molecular landscapes by means of > AFM . G. SCOLES, Princeton University, USA; Linear and Non linear Optical > Microscopy . C. SHEPPARD, Ntl Univ. of Singapore, Singapore; Optical > Microscopy, 3D imaging, Photonic Forces . E. STELZER, EMBL, Heidelberg, > DE; Laser scissors and tweezers in cell biology . I. TOLIC-NORRELYKKE, > MPI, Dresden, DE; Fluorescence imaging in Neuroscience . V. TORRE, SISSA, > Trieste, IT; Quantitative colocalization . C. USAI, CNR-IBF, Genoa, IT; > Confocal Microscopy, Structured light methods . T. WILSON, University of > Oxford, UK; Photoswitch-activatable fluorescent proteins, Lifetime . F. > WOUTERS, Univ. of Goettingen, DE. > -------- > ---------------------------------------------------- > "Follow knowledge wherever it leads us." (reading Galileo Galilei) > ----------------------------------------------------- > Alberto Diaspro, EBSA President-Elect, MicroScoBIO LAMBS-IFOM, Department > of Physics, University of Genoa, Via Dodecaneso 33, 16146 Genoa, Italy - > fax +39-010314218 - tel +39 0103536426/309; URLs: www.lambs.it; > EBSA is Biophysics in Europe - European Biophysical Societies' Association > www.ebsa.org > ---------------------------------------------- > > ________________________________________________________________________ > SERVIZIO VOICE: TELEFONA e INVIA SMS dal tuo computer a tariffe > vantaggiose! > Scopri come telefonare e videochiamare gratis da pc a pc. > http://voice.repubblica.it > |
|
S. Pagakis (IIBEAA) |
|
|
In reply to this post by Guy Cox
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of George McNamara > Sent: Sunday, 9 December 2007 6:08 AM > To: [hidden email] > Subject: Re: Coverslip thickness and correction collar ... option Ib > > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > option Ib. Measure of the glass thickness before you culture your > cells. Then, for this lens, use dishes whose coverglasses are 170 um > (or whatever thickness works best for the lens at its current > "optimal" setting). > > If one in five are the right thickness, than the cost of that dish, > for this experiment, has gone from about $2 to $10. Should take less > than a minute to measure a dish using a 20x, decent NA, dry lens. Dear George On a recent test I did, I discovered something that I had not expected. A dry objective would significantly underestimate the thickness of a glass coverslip. For a #1 coverslip, a 20x dry objective would measure it as 98microns thick, where as a 63oil objective as 138microns thick. For a #1.5 coverslip the values were 113microns and 160microns respectively. As I said, I wasn't expecting it because the 20 lens was coverslip corrected. So you have to be very careful when using dry objectives to measure anything in Z. regards ********************************* Stamatis Pagakis Ph.D. Biological Imaging Unit Biomedical Research Foundation, Academy of Athens [hidden email] > Compared to the cost of the confocal time (or of an LSM510), or of > your time to do manual image analysis, or cost of the analysis > computer and image analysis software, the extra cost is trivial. Plus, > you can sort the "rejects" and use them with lenses each thickness is > optimized for. > George McNamara, Ph.D. > University of Miami, Miller School of Medicine > Image Core > Miami, FL 33010 > [hidden email] > [hidden email] > 305-243-8436 office > http://home.earthlink.net/~pubspectra/ > http://home.earthlink.net/~geomcnamara/ > http://www.sylvester.org/health_pro/shared_resources/index.asp (see > Analytical Imaging Core Facility) > > > > > No virus found in this incoming message. > Checked by AVG Free Edition. > Version: 7.5.503 / Virus Database: 269.16.17/1177 - Release Date: > 7/12/2007 1:11 PM > > > No virus found in this outgoing message. > Checked by AVG Free Edition. > Version: 7.5.503 / Virus Database: 269.16.17/1178 - Release Date: > 8/12/2007 11:59 AM |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
The apparent thickness of a layer with a real depth of 'x' is 'x/n', where n is the RI (viewing from vacuum; otherwise n1/n2). I guess the coverslip correction doesn't correct for that; that's why your were surprised, I guess.
Zoltan
On Dec 10, 2007 1:52 PM, S. Pagakis (IIBEAA) <[hidden email]> wrote:
-- -- Zoltan Cseresnyes Facility manager, Imaging Suite Dept. of Zoology University of Cambridge Downing Street, Cambridge CB2 3EJ UK Tel.: (++44) (0)1223 769282 Fax.: (++44) (0)1223 336676 |
|
James Pawley |
|
|
In reply to this post by S. Pagakis (IIBEAA)
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >Search the CONFOCAL archive at >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >> >>From: Confocal Microscopy List >>[mailto:[hidden email]] On Behalf Of George McNamara >>Sent: Sunday, 9 December 2007 6:08 AM >>To: [hidden email] >>Subject: Re: Coverslip thickness and correction collar ... option Ib >> >>Search the CONFOCAL archive at >>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >> >>option Ib. Measure of the glass thickness before you culture your >>cells. Then, for this lens, use dishes whose coverglasses are 170 >>um (or whatever thickness works best for the lens at its current >>"optimal" setting). >> >>If one in five are the right thickness, than the cost of that dish, >>for this experiment, has gone from about $2 to $10. Should take >>less than a minute to measure a dish using a 20x, decent NA, dry >>lens. > > >Dear George > >On a recent test I did, I discovered something that I had not >expected. A dry objective would significantly underestimate the >thickness of a glass coverslip. > >For a #1 coverslip, a 20x dry objective would measure it as >98microns thick, where as a 63oil objective as 138microns thick. >For a #1.5 coverslip the values were 113microns and 160microns respectively. > >As I said, I wasn't expecting it because the 20 lens was coverslip corrected. > >So you have to be very careful when using dry objectives to measure >anything in Z. > >regards > >********************************* >Stamatis Pagakis Ph.D. >Biological Imaging Unit >Biomedical Research Foundation, Academy of Athens >[hidden email] Hi all. The dry, coverslip lens is ONLY corrected for imaging immediately next to the far side of a coverslip of the specified thickness. The image of the near side will be poorly corrected, and as you noted, in the wrong place. How much "the wrong place" depends on how well you fill the BFP because this determines the effective NA on the illumination side.. Dry lenses are even more fussy that oil lenses of the same NA about the thickness of the coverslip. Dry lenses are even more fussy that oil lenses of the same NA about the thickness of the coverslip. Dry lenses are even more fussy that oil lenses of the same NA about the thickness of the coverslip. This sad news brought to you with best wishes for the Holiday Season(translation: the time when we get to correct all the papers) Cheers, Jim P. > > > >>Compared to the cost of the confocal time (or of an LSM510), or of >>your time to do manual image analysis, or cost of the analysis >>computer and image analysis software, the extra cost is trivial. >>Plus, you can sort the "rejects" and use them with lenses each >>thickness is optimized for. > >>George McNamara, Ph.D. >>University of Miami, Miller School of Medicine >>Image Core >>Miami, FL 33010 >>[hidden email] >>[hidden email] >>305-243-8436 office >>http://home.earthlink.net/~pubspectra/ >>http://home.earthlink.net/~geomcnamara/ >>http://www.sylvester.org/health_pro/shared_resources/index.asp (see >>Analytical Imaging Core Facility) >> >> >> >> >>No virus found in this incoming message. >>Checked by AVG Free Edition. >>Version: 7.5.503 / Virus Database: 269.16.17/1177 - Release Date: >>7/12/2007 1:11 PM >> >> >>No virus found in this outgoing message. >> Checked by AVG Free Edition. >> Version: 7.5.503 / Virus Database: 269.16.17/1178 - Release Date: >>8/12/2007 11:59 AM -- **************************************** Prof. James B. Pawley, Ph. 608-263-3147 Room 223, Zoology Research Building, FAX 608-262-9083 250 N. Mills St., Madison, WI, 53706 [hidden email] "A scientist is not one who can answer questions but one who can question answers." Theodore Schick Jr., Skeptical Enquirer, 21-2:39 |
|
Craig Brideau |
|
|
In reply to this post by S. Pagakis (IIBEAA)
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Spherical aberration caused by the index mismatch (air to glass) will alter the apparent depth when making these sorts of measurements. Unless you know the exact index of refraction of the glass you cannot correct for this error when measuring thickness using a dry lens. Even a wet or oil lens can have some error if the index of the glass is not matched, but for dry lenses it is the worst as the index difference is highest. Craig > Dear George > > On a recent test I did, I discovered something that I had not expected. > A dry objective would significantly underestimate the thickness of a > glass coverslip. > > For a #1 coverslip, a 20x dry objective would measure it as 98microns > thick, where as a 63oil objective as 138microns thick. > For a #1.5 coverslip the values were 113microns and 160microns > respectively. > > As I said, I wasn't expecting it because the 20 lens was coverslip > corrected. > > So you have to be very careful when using dry objectives to measure > anything in Z. > > regards > > ********************************* > Stamatis Pagakis Ph.D. > Biological Imaging Unit > Biomedical Research Foundation, Academy of Athens > [hidden email] > > > > > > Compared to the cost of the confocal time (or of an LSM510), or of > > your time to do manual image analysis, or cost of the analysis > > computer and image analysis software, the extra cost is trivial. Plus, > > you can sort the "rejects" and use them with lenses each thickness is > > optimized for. > > > George McNamara, Ph.D. > > University of Miami, Miller School of Medicine > > Image Core > > Miami, FL 33010 > > [hidden email] > > [hidden email] > > 305-243-8436 office > > http://home.earthlink.net/~pubspectra/ > > http://home.earthlink.net/~geomcnamara/ > > http://www.sylvester.org/health_pro/shared_resources/index.asp (see > > Analytical Imaging Core Facility) > > > > > > > > > > No virus found in this incoming message. > > Checked by AVG Free Edition. > > Version: 7.5.503 / Virus Database: 269.16.17/1177 - Release Date: > > 7/12/2007 1:11 PM > > > > > > No virus found in this outgoing message. > > Checked by AVG Free Edition. > > Version: 7.5.503 / Virus Database: 269.16.17/1178 - Release Date: > > 8/12/2007 11:59 AM > |
|
George McNamara |
|
|
In reply to this post by S. Pagakis (IIBEAA)
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal The only way to know which - if either - was the correct answer is to stand the coverglass on edge, and measure that. Did you do this? I did not mention this with respect to the dishes, because that measurement would require breaking or unglueing the glass. Jim P - thanks, thanks, and thanks again, for your point about where the lenses are corrected for. At 08:52 AM 12/10/2007, you wrote: >Search the CONFOCAL archive at >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >> >>From: Confocal Microscopy List >>[mailto:[hidden email]] On Behalf Of George McNamara >>Sent: Sunday, 9 December 2007 6:08 AM >>To: [hidden email] >>Subject: Re: Coverslip thickness and correction collar ... option Ib >> >>Search the CONFOCAL archive at >>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >> >>option Ib. Measure of the glass thickness before you culture your >>cells. Then, for this lens, use dishes whose coverglasses are 170 >>um (or whatever thickness works best for the lens at its current >>"optimal" setting). >> >>If one in five are the right thickness, than the cost of that dish, >>for this experiment, has gone from about $2 to $10. Should take >>less than a minute to measure a dish using a 20x, decent NA, dry lens. > > >Dear George > >On a recent test I did, I discovered something that I had not >expected. A dry objective would significantly underestimate the >thickness of a glass coverslip. > >For a #1 coverslip, a 20x dry objective would measure it as >98microns thick, where as a 63oil objective as 138microns thick. >For a #1.5 coverslip the values were 113microns and 160microns respectively. > >As I said, I wasn't expecting it because the 20 lens was coverslip corrected. > >So you have to be very careful when using dry objectives to measure >anything in Z. > >regards > >********************************* >Stamatis Pagakis Ph.D. >Biological Imaging Unit >Biomedical Research Foundation, Academy of Athens >[hidden email] > > > > >>Compared to the cost of the confocal time (or of an LSM510), or of >>your time to do manual image analysis, or cost of the analysis >>computer and image analysis software, the extra cost is trivial. >>Plus, you can sort the "rejects" and use them with lenses each >>thickness is optimized for. > >>George McNamara, Ph.D. >>University of Miami, Miller School of Medicine >>Image Core >>Miami, FL 33010 >>[hidden email] >>[hidden email] >>305-243-8436 office >>http://home.earthlink.net/~pubspectra/ >>http://home.earthlink.net/~geomcnamara/ >>http://www.sylvester.org/health_pro/shared_resources/index.asp (see >>Analytical Imaging Core Facility) >> >> >> >> >>No virus found in this incoming message. >>Checked by AVG Free Edition. >>Version: 7.5.503 / Virus Database: 269.16.17/1177 - Release Date: >>7/12/2007 1:11 PM >> >> >>No virus found in this outgoing message. >> Checked by AVG Free Edition. >> Version: 7.5.503 / Virus Database: 269.16.17/1178 - Release Date: >> 8/12/2007 11:59 AM > > > > > > >George McNamara, Ph.D. >University of Miami, Miller School of Medicine >Image Core >Miami, FL 33010 >[hidden email] >[hidden email] >305-243-8436 office >http://home.earthlink.net/~pubspectra/ >http://home.earthlink.net/~geomcnamara/ >http://www.sylvester.org/health_pro/shared_resources/index.asp (see >Analytical Imaging Core Facility) > > |
|
Jan Peychl |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear all, as posted here already some time ago by Jens Rietdorf, there is a company which sells coverslips of selected thickness. I hope that this might be of help ( no commercial interest from my side).: http://www.hecht-assistent.de/e_microscopic/cover_glasses.html search for : Assistent" Cover glasses, selected thickness, CE 1014 thickness 0,17 ± 0,01 mm 1015 thickness 0,17 ± 0,02 mm Hope that this might help. Jan George McNamara wrote: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > The only way to know which - if either - was the correct answer is to > stand the coverglass on edge, and measure that. Did you do this? > > I did not mention this with respect to the dishes, because that > measurement would require breaking or unglueing the glass. > > Jim P - thanks, thanks, and thanks again, for your point about where > the lenses are corrected for. > > > > At 08:52 AM 12/10/2007, you wrote: > >> Search the CONFOCAL archive at >> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >> >>> >>> From: Confocal Microscopy List >>> [mailto:[hidden email]] On Behalf Of George McNamara >>> Sent: Sunday, 9 December 2007 6:08 AM >>> To: [hidden email] >>> Subject: Re: Coverslip thickness and correction collar ... option Ib >>> >>> Search the CONFOCAL archive at >>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >>> >>> option Ib. Measure of the glass thickness before you culture your >>> cells. Then, for this lens, use dishes whose coverglasses are 170 um >>> (or whatever thickness works best for the lens at its current >>> "optimal" setting). >>> >>> If one in five are the right thickness, than the cost of that dish, >>> for this experiment, has gone from about $2 to $10. Should take less >>> than a minute to measure a dish using a 20x, decent NA, dry lens. >> >> >> >> Dear George >> >> On a recent test I did, I discovered something that I had not >> expected. A dry objective would significantly underestimate the >> thickness of a glass coverslip. >> >> For a #1 coverslip, a 20x dry objective would measure it as 98microns >> thick, where as a 63oil objective as 138microns thick. >> For a #1.5 coverslip the values were 113microns and 160microns >> respectively. >> >> As I said, I wasn't expecting it because the 20 lens was coverslip >> corrected. >> >> So you have to be very careful when using dry objectives to measure >> anything in Z. >> >> regards >> >> ********************************* >> Stamatis Pagakis Ph.D. >> Biological Imaging Unit >> Biomedical Research Foundation, Academy of Athens >> [hidden email] >> >> >> >> >>> Compared to the cost of the confocal time (or of an LSM510), or of >>> your time to do manual image analysis, or cost of the analysis >>> computer and image analysis software, the extra cost is trivial. >>> Plus, you can sort the "rejects" and use them with lenses each >>> thickness is optimized for. >> >> >>> George McNamara, Ph.D. >>> University of Miami, Miller School of Medicine >>> Image Core >>> Miami, FL 33010 >>> [hidden email] >>> [hidden email] >>> 305-243-8436 office >>> http://home.earthlink.net/~pubspectra/ >>> http://home.earthlink.net/~geomcnamara/ >>> http://www.sylvester.org/health_pro/shared_resources/index.asp (see >>> Analytical Imaging Core Facility) >>> >>> >>> >>> >>> No virus found in this incoming message. >>> Checked by AVG Free Edition. >>> Version: 7.5.503 / Virus Database: 269.16.17/1177 - Release Date: >>> 7/12/2007 1:11 PM >>> >>> >>> No virus found in this outgoing message. >>> Checked by AVG Free Edition. >>> Version: 7.5.503 / Virus Database: 269.16.17/1178 - Release Date: >>> 8/12/2007 11:59 AM >> >> >> >> >> >> >> >> George McNamara, Ph.D. >> University of Miami, Miller School of Medicine >> Image Core >> Miami, FL 33010 >> [hidden email] >> [hidden email] >> 305-243-8436 office >> http://home.earthlink.net/~pubspectra/ >> http://home.earthlink.net/~geomcnamara/ >> http://www.sylvester.org/health_pro/shared_resources/index.asp (see >> Analytical Imaging Core Facility) >> >> |