DAPI color change issue

classic Classic list List threaded Threaded
♦
♦
12 messages Options
G. Esteban Fernandez G. Esteban Fernandez
Reply | Threaded
Open this post in threaded view
♦
♦
|

DAPI color change issue

Hi all,
 
One of my users came to me with immuno-stained paraffin sections of mouse cardiac muscle mounted in Vectashield + DAPI (secondary Abs were Alexa488 and Alexa568).  The problem is that the nuclei are fluorescing red under UV excitation.  Has anyone ever seen this apparent shift in emission of DAPI (e.g. with old Vectashield) or have an idea of what's going on?
 
Thanks,
Esteban

--
G. Esteban Fernandez, Ph.D.
Associate Director
Molecular Cytology Core Facility
University of Missouri
120 Bond Life Sciences Center
Columbia, MO  65211

http://www.biotech.missouri.edu/mcc/

573-882-4895
573-884-9395 fax

Mayandi Sivaguru Mayandi Sivaguru
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DAPI color change issue


Esteban, as you have predicted, it sounds like the old vectashield, have you seen any speckles also in the sample fluorescing red, then it might be phenylendiamine, the antifade used in vectashield would have precipitated (usually a brown precipitate under visible light), but I have not seen them under UV light. Using a new batch might answer the question.

However, we reccommend, using DAPI always as an aqueous incubation in PBS rather than in the mounting medium, for better penetration.
Shiv

At 04:18 PM 1/20/2009, you wrote:
Hi all,
 
One of my users came to me with immuno-stained paraffin sections of mouse cardiac muscle mounted in Vectashield + DAPI (secondary Abs were Alexa488 and Alexa568). The problem is that the nuclei are fluorescing red under UV excitation. Has anyone ever seen this apparent shift in emission of DAPI (e.g. with old Vectashield) or have an idea of what's going on?
 
Thanks,
Esteban

--
G. Esteban Fernandez, Ph.D.
Associate Director
Molecular Cytology Core Facility
University of Missouri
120 Bond Life Sciences Center
Columbia, MO 65211

http://www.biotech.missouri.edu/mcc/

573-882-4895
573-884-9395 fax

Mayandi Sivaguru, PhD, PhD
Microscopy Facility Manager
8, Institute for Genomic Biology
University of Illinois at Urbana-Champaign
1206 West Gregory Dr.
Urbana, IL 61801 USA

Office: 217.333.1214
Fax: 217.244.2496
[hidden email]
http://core.igb.uiuc.edu

Dale Callaham Dale Callaham
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DAPI color change issue

DAPI is not very soluble with phosphate. See the data sheet. They do
mention using it in PBS (at a working dilution, I assume) but don't use
PBS for a stock - big mistake.. trust me....

http://probes.invitrogen.com/media/pis/mp01306.pdf
==================================
Preparing the DAPI
Stock Solution To make a 5 mg/mL DAPI stock solution (14.3 mM for the
dihydrochloride or 10.9 mM for the dilactate), dissolve the contents of
one vial (10 mg) in 2 mL of deionized water (dH2O) or dimethylformamide
(DMF). The less water-soluble DAPI dihydro-chloride may take some time
to completely dissolve in water and sonication may be necessary.

Note: Neither of these DAPI derivatives is very soluble in
phosphate-buffered saline (PBS).
====================================

Dale


Mayandi Sivaguru wrote:

>
> Esteban, as you have predicted, it sounds like the old vectashield, have
> you seen any speckles also in the sample fluorescing red, then it might
> be phenylendiamine, the antifade used in vectashield would have
> precipitated (usually a brown precipitate under visible light), but I
> have not seen them under UV light. Using a new batch might answer the
> question.
>
> However, we reccommend, using DAPI always as an aqueous incubation in
> PBS rather than in the mounting medium, for better penetration.
> Shiv
>
> At 04:18 PM 1/20/2009, you wrote:
>> Hi all,
>>  
>> One of my users came to me with immuno-stained paraffin sections of
>> mouse cardiac muscle mounted in Vectashield + DAPI (secondary Abs were
>> Alexa488 and Alexa568). The problem is that the nuclei are fluorescing
>> red under UV excitation. Has anyone ever seen this apparent shift in
>> emission of DAPI (e.g. with old Vectashield) or have an idea of what's
>> going on?
>>  
>> Thanks,
>> Esteban
>>
>> --
>> G. Esteban Fernandez, Ph.D.
>> Associate Director
>> Molecular Cytology Core Facility
>> University of Missouri
>> 120 Bond Life Sciences Center
>> Columbia, MO 65211
>>
>> http://www.biotech.missouri.edu/mcc/
>>
>> 573-882-4895
>> 573-884-9395 fax
>
> Mayandi Sivaguru, PhD, PhD
> Microscopy Facility Manager
> 8, Institute for Genomic Biology
> University of Illinois at Urbana-Champaign
> 1206 West Gregory Dr.
> Urbana, IL 61801 USA
>
> Office: 217.333.1214
> Fax: 217.244.2496
> [hidden email]
> http://core.igb.uiuc.edu
>
> <http://core.igb.uiuc.edu/>
>
B. Prabhakar Pandian B. Prabhakar Pandian
Reply | Threaded
Open this post in threaded view
♦
♦
|

Mercury Bulbs

In reply to this post by Mayandi Sivaguru
Hello:

We have a mercury bulb lamphouse attached to our microscope. Can anybody
tell where I can find cheap bulbs for the lamphouse. These are 100W bulbs.


Thanks,

-Prabhakar
kspencer007 kspencer007
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Mercury Bulbs

Bulbtronics or Bulb Connection work for us.
        Kathy
 

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of B. Prabhakar Pandian
Sent: Wednesday, January 21, 2009 11:32 AM
To: [hidden email]
Subject: Mercury Bulbs

Hello:

We have a mercury bulb lamphouse attached to our microscope. Can anybody
tell where I can find cheap bulbs for the lamphouse. These are 100W
bulbs.


Thanks,

-Prabhakar
Theresa Swayne Theresa Swayne
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Mercury Bulbs

In reply to this post by B. Prabhakar Pandian

On Jan 21, 2009, at 2:32 PM, B. Prabhakar Pandian wrote:

We have a mercury bulb lamphouse attached to our microscope. Can anybody tell where I can find cheap bulbs for the lamphouse. These are 100W bulbs.

Fisher and VWR have bulbs too, and they work out to be pretty cheap for us because of our institutional discount. You may have something similar.

I've had good experiences with both Ushio and Osram bulbs.

Hope this helps,
Theresa

------------------------------------
Theresa C. Swayne, Ph.D.
Manager, Confocal and Specialized Microscopy Shared Resource
Herbert Irving Comprehensive Cancer Center, Columbia University
1130 St Nicholas Ave, 222A
New York, NY 10032

212-851-4613


Dale Callaham Dale Callaham
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Mercury Bulbs

In reply to this post by B. Prabhakar Pandian
http://www.bulbdirect.com/ is where I get mine (no affiliation) - they
are very prompt. The standard HBO 100W/2 is ~ $115. They have Osram and
Ushio typically. The website seems to designed by people who don't use
websites, but I have always had ecellent service.
http://www.bulbdirect.com/search/text_search.php?search=HBO+100W&Submit=Go%21

B. Prabhakar Pandian wrote:
> Hello:
>
> We have a mercury bulb lamphouse attached to our microscope. Can anybody
> tell where I can find cheap bulbs for the lamphouse. These are 100W bulbs.
>
>
> Thanks,
>
> -Prabhakar
Christian-103 Christian-103
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Mercury Bulbs

In reply to this post by Theresa Swayne
We've always used bulbdirect.com.  They do only bulbs, so the service is great.



Theresa Swayne <[hidden email]> wrote:

On Jan 21, 2009, at 2:32 PM, B. Prabhakar Pandian wrote:

We have a mercury bulb lamphouse attached to our microscope. Can anybody tell where I can find cheap bulbs for the lamphouse. These are 100W bulbs.

Fisher and VWR have bulbs too, and they work out to be pretty cheap for us because of our institutional discount. You may have something similar.

I've had good experiences with both Ushio and Osram bulbs.

Hope this helps,
Theresa

------------------------------------
Theresa C. Swayne, Ph.D.
Manager, Confocal and Specialized Microscopy Shared Resource
Herbert Irving Comprehensive Cancer Center, Columbia University
1130 St Nicholas Ave, 222A
New York, NY 10032

212-851-4613





Moninger, Thomas Moninger, Thomas
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Mercury Bulbs

We get them From Sitler’s Supplies, 800-426-3938.

 

$89 plus shipping.

 

Tom

 

Thomas Moninger [hidden email]

Assistant Director

University of Iowa Central Microscopy Research Facilities

www.uiowa.edu/~cemrf

P Think Green before printing this e-mail!


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christian
Sent: Wednesday, January 21, 2009 2:23 PM
To: [hidden email]
Subject: Re: Mercury Bulbs

 

We've always used bulbdirect.com.  They do only bulbs, so the service is great.



Theresa Swayne <[hidden email]> wrote:

 

On Jan 21, 2009, at 2:32 PM, B. Prabhakar Pandian wrote:



We have a mercury bulb lamphouse attached to our microscope. Can anybody tell where I can find cheap bulbs for the lamphouse. These are 100W bulbs.

 

Fisher and VWR have bulbs too, and they work out to be pretty cheap for us because of our institutional discount. You may have something similar.

 

I've had good experiences with both Ushio and Osram bulbs.

 

Hope this helps,

Theresa

 

------------------------------------

Theresa C. Swayne, Ph.D.

Manager, Confocal and Specialized Microscopy Shared Resource

Herbert Irving Comprehensive Cancer Center, Columbia University

1130 St Nicholas Ave, 222A

New York, NY 10032

 

212-851-4613

 

 



Chere Petty Chere Petty
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Mercury Bulbs

In reply to this post by B. Prabhakar Pandian
B. Prabhakar Pandian wrote:

> Hello:
>
> We have a mercury bulb lamphouse attached to our microscope. Can
> anybody tell where I can find cheap bulbs for the lamphouse. These are
> 100W bulbs.
>
>
> Thanks,
>
> -Prabhakar
>
I like bulbman..com

--
Chere Petty, M.S.
Manager of UMBC Keith R. Porter Core Imaging Facility
Department of Biological Sciences
University of Maryland Baltimore County (UMBC)
1000 Hilltop Circle
Baltimore, Maryland 21250
Fax: 410-455-3875
Cell: 301-367-8408
[hidden email]
www.emumbc.com
B. Prabhakar Pandian B. Prabhakar Pandian
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Mercury Bulbs

Thanks to everybody for providing me the link to their bulb shop.

Regards,

-Prabhakar
Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Mercury Bulbs

In reply to this post by B. Prabhakar Pandian
You don't say where you are, which makes it a bit difficult.
I can say that in general I've found it cheaper to buy from
specialist lamp suppliers than from microscope companies.

                                                 Guy

Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net
-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of B. Prabhakar Pandian
Sent: Thursday, 22 January 2009 6:32 AM
To: [hidden email]
Subject: Mercury Bulbs

Hello:

We have a mercury bulb lamphouse attached to our microscope. Can anybody tell where I can find cheap bulbs for the lamphouse. These are 100W bulbs.


Thanks,

-Prabhakar