Hi all,
One of my users came to me with immuno-stained paraffin sections of mouse cardiac muscle mounted in Vectashield + DAPI (secondary Abs were Alexa488 and Alexa568). The problem is that the nuclei are fluorescing red under UV excitation. Has anyone ever seen this apparent shift in emission of DAPI (e.g. with old Vectashield) or have an idea of what's going on?
Thanks,
Esteban
-- |
Esteban, as you have predicted, it sounds like the old vectashield, have you seen any speckles also in the sample fluorescing red, then it might be phenylendiamine, the antifade used in vectashield would have precipitated (usually a brown precipitate under visible light), but I have not seen them under UV light. Using a new batch might answer the question. However, we reccommend, using DAPI always as an aqueous incubation in PBS rather than in the mounting medium, for better penetration. Shiv At 04:18 PM 1/20/2009, you wrote: Hi all, Microscopy Facility Manager 8, Institute for Genomic Biology University of Illinois at Urbana-Champaign 1206 West Gregory Dr. Urbana, IL 61801 USA Office: 217.333.1214 Fax: 217.244.2496 [hidden email] http://core.igb.uiuc.edu |
DAPI is not very soluble with phosphate. See the data sheet. They do
mention using it in PBS (at a working dilution, I assume) but don't use PBS for a stock - big mistake.. trust me.... http://probes.invitrogen.com/media/pis/mp01306.pdf ================================== Preparing the DAPI Stock Solution To make a 5 mg/mL DAPI stock solution (14.3 mM for the dihydrochloride or 10.9 mM for the dilactate), dissolve the contents of one vial (10 mg) in 2 mL of deionized water (dH2O) or dimethylformamide (DMF). The less water-soluble DAPI dihydro-chloride may take some time to completely dissolve in water and sonication may be necessary. Note: Neither of these DAPI derivatives is very soluble in phosphate-buffered saline (PBS). ==================================== Dale Mayandi Sivaguru wrote: > > Esteban, as you have predicted, it sounds like the old vectashield, have > you seen any speckles also in the sample fluorescing red, then it might > be phenylendiamine, the antifade used in vectashield would have > precipitated (usually a brown precipitate under visible light), but I > have not seen them under UV light. Using a new batch might answer the > question. > > However, we reccommend, using DAPI always as an aqueous incubation in > PBS rather than in the mounting medium, for better penetration. > Shiv > > At 04:18 PM 1/20/2009, you wrote: >> Hi all, >> >> One of my users came to me with immuno-stained paraffin sections of >> mouse cardiac muscle mounted in Vectashield + DAPI (secondary Abs were >> Alexa488 and Alexa568). The problem is that the nuclei are fluorescing >> red under UV excitation. Has anyone ever seen this apparent shift in >> emission of DAPI (e.g. with old Vectashield) or have an idea of what's >> going on? >> >> Thanks, >> Esteban >> >> -- >> G. Esteban Fernandez, Ph.D. >> Associate Director >> Molecular Cytology Core Facility >> University of Missouri >> 120 Bond Life Sciences Center >> Columbia, MO 65211 >> >> http://www.biotech.missouri.edu/mcc/ >> >> 573-882-4895 >> 573-884-9395 fax > > Mayandi Sivaguru, PhD, PhD > Microscopy Facility Manager > 8, Institute for Genomic Biology > University of Illinois at Urbana-Champaign > 1206 West Gregory Dr. > Urbana, IL 61801 USA > > Office: 217.333.1214 > Fax: 217.244.2496 > [hidden email] > http://core.igb.uiuc.edu > > <http://core.igb.uiuc.edu/> > |
In reply to this post by Mayandi Sivaguru
Hello:
We have a mercury bulb lamphouse attached to our microscope. Can anybody tell where I can find cheap bulbs for the lamphouse. These are 100W bulbs. Thanks, -Prabhakar |
Bulbtronics or Bulb Connection work for us.
Kathy -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of B. Prabhakar Pandian Sent: Wednesday, January 21, 2009 11:32 AM To: [hidden email] Subject: Mercury Bulbs Hello: We have a mercury bulb lamphouse attached to our microscope. Can anybody tell where I can find cheap bulbs for the lamphouse. These are 100W bulbs. Thanks, -Prabhakar |
In reply to this post by B. Prabhakar Pandian
On Jan 21, 2009, at 2:32 PM, B. Prabhakar Pandian wrote:
Fisher and VWR have bulbs too, and they work out to be pretty cheap for us because of our institutional discount. You may have something similar. I've had good experiences with both Ushio and Osram bulbs. Hope this helps, Theresa ------------------------------------ Theresa C. Swayne, Ph.D. Manager, Confocal and Specialized Microscopy Shared Resource Herbert Irving Comprehensive Cancer Center, Columbia University 1130 St Nicholas Ave, 222A New York, NY 10032 212-851-4613 |
In reply to this post by B. Prabhakar Pandian
http://www.bulbdirect.com/ is where I get mine (no affiliation) - they
are very prompt. The standard HBO 100W/2 is ~ $115. They have Osram and Ushio typically. The website seems to designed by people who don't use websites, but I have always had ecellent service. http://www.bulbdirect.com/search/text_search.php?search=HBO+100W&Submit=Go%21 B. Prabhakar Pandian wrote: > Hello: > > We have a mercury bulb lamphouse attached to our microscope. Can anybody > tell where I can find cheap bulbs for the lamphouse. These are 100W bulbs. > > > Thanks, > > -Prabhakar |
In reply to this post by Theresa Swayne
We've always used bulbdirect.com. They do only bulbs, so the service is great.
Theresa Swayne <[hidden email]> wrote:
|
We get them From Sitler’s Supplies, 800-426-3938.
$89 plus shipping. Tom Thomas Moninger [hidden email] Assistant Director P Think
Green before printing this e-mail! From: Confocal
Microscopy List [mailto:[hidden email]] On Behalf Of Christian We've always used bulbdirect.com. They do only bulbs, so the
service is great. On Jan 21, 2009, at 2:32 PM, B. Prabhakar Pandian wrote:
We have a mercury bulb
lamphouse attached to our microscope. Can anybody tell where I can find cheap
bulbs for the lamphouse. These are 100W bulbs. Fisher and VWR have
bulbs too, and they work out to be pretty cheap for us because of our
institutional discount. You may have something similar. I've had good
experiences with both Ushio and Osram bulbs. Hope this helps, Theresa ------------------------------------ Theresa C. Swayne, Ph.D. Manager, Confocal and Specialized Microscopy
Shared Resource 212-851-4613
|
In reply to this post by B. Prabhakar Pandian
B. Prabhakar Pandian wrote:
> Hello: > > We have a mercury bulb lamphouse attached to our microscope. Can > anybody tell where I can find cheap bulbs for the lamphouse. These are > 100W bulbs. > > > Thanks, > > -Prabhakar > -- Chere Petty, M.S. Manager of UMBC Keith R. Porter Core Imaging Facility Department of Biological Sciences University of Maryland Baltimore County (UMBC) 1000 Hilltop Circle Baltimore, Maryland 21250 Fax: 410-455-3875 Cell: 301-367-8408 [hidden email] www.emumbc.com |
Thanks to everybody for providing me the link to their bulb shop.
Regards, -Prabhakar |
In reply to this post by B. Prabhakar Pandian
You don't say where you are, which makes it a bit difficult.
I can say that in general I've found it cheaper to buy from specialist lamp suppliers than from microscope companies. Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________ Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of B. Prabhakar Pandian Sent: Thursday, 22 January 2009 6:32 AM To: [hidden email] Subject: Mercury Bulbs Hello: We have a mercury bulb lamphouse attached to our microscope. Can anybody tell where I can find cheap bulbs for the lamphouse. These are 100W bulbs. Thanks, -Prabhakar No virus found in this incoming message. Checked by AVG. Version: 7.5.552 / Virus Database: 270.10.10/1905 - Release Date: 20/01/2009 2:34 PM No virus found in this outgoing message. Checked by AVG. Version: 7.5.552 / Virus Database: 270.10.10/1906 - Release Date: 21/01/2009 7:07 AM |
Free forum by Nabble | Edit this page |