Dapi stain emits at 488,

classic Classic list List threaded Threaded
♦
♦
7 messages Options
Alejandro Roth-2 Alejandro Roth-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Dapi stain emits at 488,

Hello
I know this is not strictly a confocal question but, we are staining with
DAPI and alexa-488 and get a VERY strange bleed through.
1. We illuminate with 488 and observe cytoplasmic staining (nothing in the
nucleus).
2. We then illuminate with UV light from the HBO and observe a nice DAPI stain.
3. If we then re-illuminate with the 488 laser, the nucleus is now lit!
(these are PFA fixed cortical neuron primary cultures).
We have diluted DAPI to the point of loosing the signal, and this still occurs.

Any suggestions? Any explication on what is going on?

Thank you all.
Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Dapi stain emits at 488,

Try it without any DAPI - if it still occurs it must be
autofluorescence (probably from the pfa) which is being
photoactivated by the UV light.  If so, try methanol fixation
instead and see if it goes away.

                                                  Guy

Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
 

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of Alejandro Roth
Sent: Friday, 5 December 2008 6:24 AM
To: [hidden email]
Subject: Dapi stain emits at 488,

Hello
I know this is not strictly a confocal question but, we are staining
with
DAPI and alexa-488 and get a VERY strange bleed through.
1. We illuminate with 488 and observe cytoplasmic staining (nothing in
the
nucleus).
2. We then illuminate with UV light from the HBO and observe a nice DAPI
stain.
3. If we then re-illuminate with the 488 laser, the nucleus is now lit!
(these are PFA fixed cortical neuron primary cultures).
We have diluted DAPI to the point of loosing the signal, and this still
occurs.

Any suggestions? Any explication on what is going on?

Thank you all.
leoncio vergara leoncio vergara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Dapi stain emits at 488,

I cannot offer an explanation but I can corroborate we have also seen the same phenomenon in wide field epifluorescence. We never investigated the problem further, in our case it was enough to just get the DAPI image last and work around the problem.


Leoncio A. Vergara MD

Assistant Professor
Laboratory of Protein Misfolding Diseases (lab-PMD),
George and Cynthia Mitchell Center for Neurodegenerative Diseases Research.

Director of the Optical Imaging Lab. (OIL),

Dept. of Neuroscience and Cell Biology

University of Texas Medical Branch (UTMB)

301 University Blvd

Galveston, Texas 77555-0641

OIL phone: 409-772-3970  

Lab-PMD phone: 409-7470019

fax: 409-7470015


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Guy Cox
Sent: Thursday, December 04, 2008 7:00 PM
To: [hidden email]
Subject: Re: Dapi stain emits at 488,

Try it without any DAPI - if it still occurs it must be autofluorescence (probably from the pfa) which is being photoactivated by the UV light.  If so, try methanol fixation instead and see if it goes away.

                                                  Guy

Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
 

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of Alejandro Roth
Sent: Friday, 5 December 2008 6:24 AM
To: [hidden email]
Subject: Dapi stain emits at 488,

Hello
I know this is not strictly a confocal question but, we are staining
with
DAPI and alexa-488 and get a VERY strange bleed through.
1. We illuminate with 488 and observe cytoplasmic staining (nothing in
the
nucleus).
2. We then illuminate with UV light from the HBO and observe a nice DAPI
stain.
3. If we then re-illuminate with the 488 laser, the nucleus is now lit!
(these are PFA fixed cortical neuron primary cultures).
We have diluted DAPI to the point of loosing the signal, and this still
occurs.

Any suggestions? Any explication on what is going on?

Thank you all.
McDonald, David L McDonald, David L
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Dapi stain emits at 488,

In reply to this post by Alejandro Roth-2
I have seen the same phenomenon in our Molecular Probes FluoCells
Slide (#1) using our widefield scopes.  Strangely, it seems to occur
(or is at least much more obvious) only in certain areas of the slide.

Dave


Dave McDonald
Scientific Imaging Lab
Fred Hutchinson Cancer Research Center
1100 Fairview Avenue North, DE-512
Seattle, WA 98109
206-667-4205
http://www.fhcrc.org


At 11:23 AM 12/4/2008, you wrote:

>Hello
>I know this is not strictly a confocal question but, we are staining with
>DAPI and alexa-488 and get a VERY strange bleed through.
>1. We illuminate with 488 and observe cytoplasmic staining (nothing in the
>nucleus).
>2. We then illuminate with UV light from the HBO and observe a nice
>DAPI stain.
>3. If we then re-illuminate with the 488 laser, the nucleus is now lit!
>(these are PFA fixed cortical neuron primary cultures).
>We have diluted DAPI to the point of loosing the signal, and this
>still occurs.
>
>Any suggestions? Any explication on what is going on?
>
>Thank you all.
Brotchie, Daniel Brotchie, Daniel
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Dapi stain emits at 488,

Hi,

By chance I came across the same question today when searching the 2004 Archive.
Lots of useful comments. One subscriber - Xuejun - suggested it was
photoactivation of DAPI causing the nuclei to emit green light when previously
non had been detected. Xuejun solved the problem by reducing DAPI concentration
and using NDFs:

http://lists.umn.edu/cgi-bin/wa?A2=ind0406&L=confocalmicroscopy&D=0&P=6493

Daniel
------------------------------------------------
Daniel Brotchie
Ophthalmology
School of Clinical Sciences
University Clinical Departments
The Duncan Building
Daulby Street
LIVERPOOL
L69 3GA
UK
 
Tel: +44 (0) 151 706 4017
Fax: +44 (0) 151 706 5934
E-mail [hidden email]

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Dave McDonald
Sent: 05 December 2008 15:34
To: [hidden email]
Subject: Re: Dapi stain emits at 488,

I have seen the same phenomenon in our Molecular Probes FluoCells
Slide (#1) using our widefield scopes.  Strangely, it seems to occur
(or is at least much more obvious) only in certain areas of the slide.

Dave


Dave McDonald
Scientific Imaging Lab
Fred Hutchinson Cancer Research Center
1100 Fairview Avenue North, DE-512
Seattle, WA 98109
206-667-4205
http://www.fhcrc.org


At 11:23 AM 12/4/2008, you wrote:

>Hello
>I know this is not strictly a confocal question but, we are staining with
>DAPI and alexa-488 and get a VERY strange bleed through.
>1. We illuminate with 488 and observe cytoplasmic staining (nothing in the
>nucleus).
>2. We then illuminate with UV light from the HBO and observe a nice
>DAPI stain.
>3. If we then re-illuminate with the 488 laser, the nucleus is now lit!
>(these are PFA fixed cortical neuron primary cultures).
>We have diluted DAPI to the point of loosing the signal, and this
>still occurs.
>
>Any suggestions? Any explication on what is going on?
>
>Thank you all.
Stephen Cody-2 Stephen Cody-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Dapi stain emits at 488,

We have seen this too. I favour the photoconversion theory. Just always image DAPI last and you will not have a problem.
 
Cheers
Stephen Cody

2008/12/6 Brotchie, Daniel <[hidden email]>
Hi,

By chance I came across the same question today when searching the 2004 Archive.
Lots of useful comments. One subscriber - Xuejun - suggested it was
photoactivation of DAPI causing the nuclei to emit green light when previously
non had been detected. Xuejun solved the problem by reducing DAPI concentration
and using NDFs:

http://lists.umn.edu/cgi-bin/wa?A2=ind0406&L=confocalmicroscopy&D=0&P=6493

Daniel
------------------------------------------------
Daniel Brotchie
Ophthalmology
School of Clinical Sciences
University Clinical Departments
The Duncan Building
Daulby Street
LIVERPOOL
L69 3GA
UK

Tel: +44 (0) 151 706 4017
Fax: +44 (0) 151 706 5934
E-mail [hidden email]

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Dave McDonald
Sent: 05 December 2008 15:34
To: [hidden email]
Subject: Re: Dapi stain emits at 488,

I have seen the same phenomenon in our Molecular Probes FluoCells
Slide (#1) using our widefield scopes.  Strangely, it seems to occur
(or is at least much more obvious) only in certain areas of the slide.

Dave


Dave McDonald
Scientific Imaging Lab
Fred Hutchinson Cancer Research Center
1100 Fairview Avenue North, DE-512
Seattle, WA 98109
206-667-4205
http://www.fhcrc.org


At 11:23 AM 12/4/2008, you wrote:
>Hello
>I know this is not strictly a confocal question but, we are staining with
>DAPI and alexa-488 and get a VERY strange bleed through.
>1. We illuminate with 488 and observe cytoplasmic staining (nothing in the
>nucleus).
>2. We then illuminate with UV light from the HBO and observe a nice
>DAPI stain.
>3. If we then re-illuminate with the 488 laser, the nucleus is now lit!
>(these are PFA fixed cortical neuron primary cultures).
>We have diluted DAPI to the point of loosing the signal, and this
>still occurs.
>
>Any suggestions? Any explication on what is going on?
>
>Thank you all.



--
Stephen Cody
Tommaso Mello Tommaso Mello
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Dapi stain emits at 488,

In reply to this post by leoncio vergara
Hi,

I also had the same problem in widefield epifluorescence. With a long-pass filter you could clearly see the DAPI changing color from bright blue to greenish even after short exposures. In my case the user was mounting her slides in PBS/glycerol without any antifade agent. We solved the issue using very diluted DAPI and mounting in ProLong (I assume other antifade mediums will work as well).

Cheers,

Tommaso


--

Tommaso Mello, PhD.
Gastroenterology Unit
Dept. of Clinical Pathophysiology 
University of Florence, Italy
Phone:  +39.055.4271-294
FAX:    +39.055.4271-297



Il giorno gio, 04/12/2008 alle 19.05 -0600, Vergara, Leoncio A. ha scritto:
I cannot offer an explanation but I can corroborate we have also seen the same phenomenon in wide field epifluorescence. We never investigated the problem further, in our case it was enough to just get the DAPI image last and work around the problem. 


Leoncio A. Vergara MD

Assistant Professor
Laboratory of Protein Misfolding Diseases (lab-PMD),
George and Cynthia Mitchell Center for Neurodegenerative Diseases Research.

Director of the Optical Imaging Lab. (OIL),

Dept. of Neuroscience and Cell Biology

University of Texas Medical Branch (UTMB)

301 University Blvd

Galveston, Texas 77555-0641

OIL phone: 409-772-3970  

Lab-PMD phone: 409-7470019 

fax: 409-7470015


-----Original Message-----
From: Confocal Microscopy List [[hidden email]] On Behalf Of Guy Cox
Sent: Thursday, December 04, 2008 7:00 PM
To: [hidden email]
Subject: Re: Dapi stain emits at 488,

Try it without any DAPI - if it still occurs it must be autofluorescence (probably from the pfa) which is being photoactivated by the UV light.  If so, try methanol fixation instead and see if it goes away.

                                                  Guy

Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
 

-----Original Message-----
From: Confocal Microscopy List [[hidden email]]
On Behalf Of Alejandro Roth
Sent: Friday, 5 December 2008 6:24 AM
To: [hidden email]
Subject: Dapi stain emits at 488,

Hello
I know this is not strictly a confocal question but, we are staining
with
DAPI and alexa-488 and get a VERY strange bleed through.
1. We illuminate with 488 and observe cytoplasmic staining (nothing in
the
nucleus).
2. We then illuminate with UV light from the HBO and observe a nice DAPI
stain.
3. If we then re-illuminate with the 488 laser, the nucleus is now lit!
(these are PFA fixed cortical neuron primary cultures).
We have diluted DAPI to the point of loosing the signal, and this still
occurs. 

Any suggestions? Any explication on what is going on?

Thank you all.