Deconvolution advice

classic Classic list List threaded Threaded
♦
♦
12 messages Options
Andrew York Andrew York
Reply | Threaded
Open this post in threaded view
♦
♦
|

Deconvolution advice

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hello, I'm looking for advice and information about deconvolution,
especially from those with first-hand experience.

Traditionally, one of the processing steps in structured illumination
microscopy is deconvolution. For our SIM, we decided to use an open-source
solution:
https://sites.google.com/site/piotrwendykier/software/deconvolution/paralleliterativedeconvolution

This seemed like a nice tradeoff between reinventing the wheel with our own
deconvolution code, and subjecting ourselves to a 'black box' closed-source
solution. However, we've recently tried out the Huygens deconvolution
software, and the results seem quite promising, possibly an improvement
over other methods we've tried. I like good images, but I don't like black
boxes, and I like to understand my data processing.

1. Is the exact algorithm used in Huygens transparently documented
anywhere? I spent a few hours searching today, but if it's out there, I
missed it.

2. Is there a clear winner for deconvolution algorithms? What should I be
using?

3. Are there other deconvolution software packages I should consider?
Ideally I'm looking for software based on clearly-documented algorithms.

Thanks for the help.

-Andrew York
NIH/NIBIB
JOEL B. SHEFFIELD JOEL B. SHEFFIELD
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Deconvolution advice

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I have found the plugins by Bob Dougherty for ImageJ to be easy to use, and
well thought-out.  I don't know too much about the exact algorithms, but
they might be worth a try.

Joel Sheffield


On Fri, Sep 28, 2012 at 4:33 PM, Andrew York <
[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hello, I'm looking for advice and information about deconvolution,
> especially from those with first-hand experience.
>
> Traditionally, one of the processing steps in structured illumination
> microscopy is deconvolution. For our SIM, we decided to use an open-source
> solution:
>
> https://sites.google.com/site/piotrwendykier/software/deconvolution/paralleliterativedeconvolution
>
> This seemed like a nice tradeoff between reinventing the wheel with our own
> deconvolution code, and subjecting ourselves to a 'black box' closed-source
> solution. However, we've recently tried out the Huygens deconvolution
> software, and the results seem quite promising, possibly an improvement
> over other methods we've tried. I like good images, but I don't like black
> boxes, and I like to understand my data processing.
>
> 1. Is the exact algorithm used in Huygens transparently documented
> anywhere? I spent a few hours searching today, but if it's out there, I
> missed it.
>
> 2. Is there a clear winner for deconvolution algorithms? What should I be
> using?
>
> 3. Are there other deconvolution software packages I should consider?
> Ideally I'm looking for software based on clearly-documented algorithms.
>
> Thanks for the help.
>
> -Andrew York
> NIH/NIBIB
>



--


Joel B. Sheffield, Ph.D
Department of Biology
Temple University
Philadelphia, PA 19122
Voice: 215 204 8839
e-mail: [hidden email]
URL:  http://astro.temple.edu/~jbs
lechristophe lechristophe
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Deconvolution advice

In reply to this post by Andrew York
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Andrew,

I don't have first-hand experience but I remember reading this two-part
comparison of different deconvolution softwares, including commercial and
open-source ones, in GIT Imaging & Microscopy by the Sage's group at EPFL
(direct links to pdfs):

http://bigwww.epfl.ch/publications/griffa1001.pdf
http://bigwww.epfl.ch/publications/griffa1002.pdf

Hope this helps,

Christophe

On Fri, Sep 28, 2012 at 10:33 PM, Andrew York <
[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hello, I'm looking for advice and information about deconvolution,
> especially from those with first-hand experience.
>
> Traditionally, one of the processing steps in structured illumination
> microscopy is deconvolution. For our SIM, we decided to use an open-source
> solution:
>
> https://sites.google.com/site/piotrwendykier/software/deconvolution/paralleliterativedeconvolution
>
> This seemed like a nice tradeoff between reinventing the wheel with our own
> deconvolution code, and subjecting ourselves to a 'black box' closed-source
> solution. However, we've recently tried out the Huygens deconvolution
> software, and the results seem quite promising, possibly an improvement
> over other methods we've tried. I like good images, but I don't like black
> boxes, and I like to understand my data processing.
>
> 1. Is the exact algorithm used in Huygens transparently documented
> anywhere? I spent a few hours searching today, but if it's out there, I
> missed it.
>
> 2. Is there a clear winner for deconvolution algorithms? What should I be
> using?
>
> 3. Are there other deconvolution software packages I should consider?
> Ideally I'm looking for software based on clearly-documented algorithms.
>
> Thanks for the help.
>
> -Andrew York
> NIH/NIBIB
>
Glen MacDonald-2 Glen MacDonald-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Deconvolution advice

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Model, etal, Journal of Microscopy, Vol. 241, Pt 1 2011, pp. 94–100
another review of decon methods and influence of parameters.

I believe the Huygens algorithm has been published.  Just email them, [hidden email].

Regards,
glen

Glen MacDonald
Core for Communication Research
Virginia Merrill Bloedel Hearing Research Center
Box 357923
University of Washington
Seattle, WA 98195-7923  USA
(206) 616-4156
[hidden email]








On Sep 28, 2012, at 2:29 PM, Christophe Leterrier wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi Andrew,
>
> I don't have first-hand experience but I remember reading this two-part
> comparison of different deconvolution softwares, including commercial and
> open-source ones, in GIT Imaging & Microscopy by the Sage's group at EPFL
> (direct links to pdfs):
>
> http://bigwww.epfl.ch/publications/griffa1001.pdf
> http://bigwww.epfl.ch/publications/griffa1002.pdf
>
> Hope this helps,
>
> Christophe
>
> On Fri, Sep 28, 2012 at 10:33 PM, Andrew York <
> [hidden email]> wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hello, I'm looking for advice and information about deconvolution,
>> especially from those with first-hand experience.
>>
>> Traditionally, one of the processing steps in structured illumination
>> microscopy is deconvolution. For our SIM, we decided to use an open-source
>> solution:
>>
>> https://sites.google.com/site/piotrwendykier/software/deconvolution/paralleliterativedeconvolution
>>
>> This seemed like a nice tradeoff between reinventing the wheel with our own
>> deconvolution code, and subjecting ourselves to a 'black box' closed-source
>> solution. However, we've recently tried out the Huygens deconvolution
>> software, and the results seem quite promising, possibly an improvement
>> over other methods we've tried. I like good images, but I don't like black
>> boxes, and I like to understand my data processing.
>>
>> 1. Is the exact algorithm used in Huygens transparently documented
>> anywhere? I spent a few hours searching today, but if it's out there, I
>> missed it.
>>
>> 2. Is there a clear winner for deconvolution algorithms? What should I be
>> using?
>>
>> 3. Are there other deconvolution software packages I should consider?
>> Ideally I'm looking for software based on clearly-documented algorithms.
>>
>> Thanks for the help.
>>
>> -Andrew York
>> NIH/NIBIB
>>
Gens, John Scott Gens, John Scott
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Deconvolution advice

In reply to this post by Andrew York
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Andrew-

You might want to get in touch with Jim McNally.  Last I heard he was
at NIH-NCI.

Some of his older papers on deconvoltion algorithms are below, but he
can probably point
you towards more recent information.

http://www.ncbi.nlm.nih.gov/pubmed/10579932

http://www.ncbi.nlm.nih.gov/pubmed/11541650   ( in particular, fig.2  
compared  a 3D image processed by three different algorithms)


Quoting Andrew York <[hidden email]>:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hello, I'm looking for advice and information about deconvolution,
> especially from those with first-hand experience.
>
> Traditionally, one of the processing steps in structured illumination
> microscopy is deconvolution. For our SIM, we decided to use an open-source
> solution:
> https://sites.google.com/site/piotrwendykier/software/deconvolution/paralleliterativedeconvolution
>
> This seemed like a nice tradeoff between reinventing the wheel with our own
> deconvolution code, and subjecting ourselves to a 'black box' closed-source
> solution. However, we've recently tried out the Huygens deconvolution
> software, and the results seem quite promising, possibly an improvement
> over other methods we've tried. I like good images, but I don't like black
> boxes, and I like to understand my data processing.
>
> 1. Is the exact algorithm used in Huygens transparently documented
> anywhere? I spent a few hours searching today, but if it's out there, I
> missed it.
>
> 2. Is there a clear winner for deconvolution algorithms? What should I be
> using?
>
> 3. Are there other deconvolution software packages I should consider?
> Ideally I'm looking for software based on clearly-documented algorithms.
>
> Thanks for the help.
>
> -Andrew York
> NIH/NIBIB
>
Gitta Hamel Gitta Hamel
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Deconvolution advice

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

**commercial response**


Hello Andrew,

It's fully understandable that people want to know the scientific
grounds when using Huygens.

For the full list of articles I refer to
http://www.svi.nl/HuygensReferences at which the relevant papers are at
the bottom of the page and mostly written during the years 1996-1998.
There are much more articles that ought to be included so your question
shows that we must give more attention to this topic.

With best wishes,

Gitta Hamel

****************************************
Gitta Hamel
Managing Director Scientific Volume Imaging bv
Developers of the *HUYGENS* software
The Netherlands
phone: ++ 31 35 6 42 16 26
*****************************************


^SVI Customer support: mail us your questions [hidden email]
<mailto:[hidden email]>or find answers online in our Huygens
WIKI:www.svi.nl/FrontPage <http://%20www.svi.nl/FrontPage>



On 09/29/2012 04:00 PM, Gens, John Scott wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Andrew-
>
> You might want to get in touch with Jim McNally.  Last I heard he was
> at NIH-NCI.
>
> Some of his older papers on deconvoltion algorithms are below, but he
> can probably point
> you towards more recent information.
>
> http://www.ncbi.nlm.nih.gov/pubmed/10579932
>
> http://www.ncbi.nlm.nih.gov/pubmed/11541650   ( in particular, fig.2  
> compared  a 3D image processed by three different algorithms)
>
>
> Quoting Andrew York <[hidden email]>:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hello, I'm looking for advice and information about deconvolution,
>> especially from those with first-hand experience.
>>
>> Traditionally, one of the processing steps in structured illumination
>> microscopy is deconvolution. For our SIM, we decided to use an
>> open-source
>> solution:
>> https://sites.google.com/site/piotrwendykier/software/deconvolution/paralleliterativedeconvolution 
>>
>>
>> This seemed like a nice tradeoff between reinventing the wheel with
>> our own
>> deconvolution code, and subjecting ourselves to a 'black box'
>> closed-source
>> solution. However, we've recently tried out the Huygens deconvolution
>> software, and the results seem quite promising, possibly an improvement
>> over other methods we've tried. I like good images, but I don't like
>> black
>> boxes, and I like to understand my data processing.
>>
>> 1. Is the exact algorithm used in Huygens transparently documented
>> anywhere? I spent a few hours searching today, but if it's out there, I
>> missed it.
>>
>> 2. Is there a clear winner for deconvolution algorithms? What should
>> I be
>> using?
>>
>> 3. Are there other deconvolution software packages I should consider?
>> Ideally I'm looking for software based on clearly-documented algorithms.
>>
>> Thanks for the help.
>>
>> -Andrew York
>> NIH/NIBIB
>>
>

--
Managing Director

Huygens SVI

tel: +31 (0)35 642 16 26

fax:  +31 (0)35 683 79 71

skype: gittahamel

cell: +31(0)618 021272

Visiting address

Laapersveld 63,
1213 VB Hilversum,
The Netherlands
George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Deconvolution advice

In reply to this post by Andrew York
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Andrew,

Wang & Smith 2012 PLoS Comput Biology,
http://www.ploscompbiol.org/article/info%3Adoi%2F10.1371%2Fjournal.pcbi.1002671

used Richardson-Lucy built into MatLab. Had some nice results though did
not deconvolve the competing methods, and botched the STED pixel size.

George




On 9/28/2012 4:33 PM, Andrew York wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hello, I'm looking for advice and information about deconvolution,
> especially from those with first-hand experience.
>
> Traditionally, one of the processing steps in structured illumination
> microscopy is deconvolution. For our SIM, we decided to use an open-source
> solution:
> https://sites.google.com/site/piotrwendykier/software/deconvolution/paralleliterativedeconvolution
>
> This seemed like a nice tradeoff between reinventing the wheel with our own
> deconvolution code, and subjecting ourselves to a 'black box' closed-source
> solution. However, we've recently tried out the Huygens deconvolution
> software, and the results seem quite promising, possibly an improvement
> over other methods we've tried. I like good images, but I don't like black
> boxes, and I like to understand my data processing.
>
> 1. Is the exact algorithm used in Huygens transparently documented
> anywhere? I spent a few hours searching today, but if it's out there, I
> missed it.
>
> 2. Is there a clear winner for deconvolution algorithms? What should I be
> using?
>
> 3. Are there other deconvolution software packages I should consider?
> Ideally I'm looking for software based on clearly-documented algorithms.
>
> Thanks for the help.
>
> -Andrew York
> NIH/NIBIB
>
>    
Unruh, Jay Unruh, Jay
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Deconvolution advice

In reply to this post by Gitta Hamel
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi all,

Deconvolution for SIM is a very different story than for other techniques.  SIM by default uses an inverse filter in its reconstruction to recombine the shifted components of the fourier transform without enhancing the high frequency noise.  Typical SIM software has a noise parameter for the wiener filter.  If you set this filter low, you start to see a rippling pattern in the noise and eventually in the high signal regions.  As far as I can tell, no one has tried to use more advanced (poisson noise driven) algorithms for this problem.  

If one assumes that Autoquant is using some variant of the algorithm shown in Tim Holmes' Handbook of Biological Confocal  chapter, then this algorithm is not immediately applicable to the SIM problem.  The algorithm update function involves dividing the original image by the convolved object guess and then convolving that ratio with the reflected psf and finally multiplying by the object guess.  Given that the raw SIM image is actually a frequency modulated image in a particular direction, it is not clear how the original ratio would be generated.  Richardson-Lucy has a similar problem.  Would you deconvolve before reconstructing?  This is the only circumstance under which the noise could be considered poisson.  In that case, can the original PSF be used?  I'm not entirely certain that the deconvolution would preserve the frequency modulation in the image.  In addition, the original reconstruction algorithm would require a second step of deconvolution during the reconstruction step--not sure how the noise parameter should be chosen after initial deconvolution.

Jay

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Gitta Hamel
Sent: Monday, October 01, 2012 8:42 AM
To: [hidden email]
Subject: Re: Deconvolution advice

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

**commercial response**


Hello Andrew,

It's fully understandable that people want to know the scientific grounds when using Huygens.

For the full list of articles I refer to http://www.svi.nl/HuygensReferences at which the relevant papers are at the bottom of the page and mostly written during the years 1996-1998.
There are much more articles that ought to be included so your question shows that we must give more attention to this topic.

With best wishes,

Gitta Hamel

****************************************
Gitta Hamel
Managing Director Scientific Volume Imaging bv Developers of the *HUYGENS* software The Netherlands
phone: ++ 31 35 6 42 16 26
*****************************************


^SVI Customer support: mail us your questions [hidden email] <mailto:[hidden email]>or find answers online in our Huygens WIKI:www.svi.nl/FrontPage <http://%20www.svi.nl/FrontPage>



On 09/29/2012 04:00 PM, Gens, John Scott wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Andrew-
>
> You might want to get in touch with Jim McNally.  Last I heard he was
> at NIH-NCI.
>
> Some of his older papers on deconvoltion algorithms are below, but he
> can probably point you towards more recent information.
>
> http://www.ncbi.nlm.nih.gov/pubmed/10579932
>
> http://www.ncbi.nlm.nih.gov/pubmed/11541650   ( in particular, fig.2  
> compared  a 3D image processed by three different algorithms)
>
>
> Quoting Andrew York <[hidden email]>:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hello, I'm looking for advice and information about deconvolution,
>> especially from those with first-hand experience.
>>
>> Traditionally, one of the processing steps in structured illumination
>> microscopy is deconvolution. For our SIM, we decided to use an
>> open-source
>> solution:
>> https://sites.google.com/site/piotrwendykier/software/deconvolution/p
>> aralleliterativedeconvolution
>>
>>
>> This seemed like a nice tradeoff between reinventing the wheel with
>> our own deconvolution code, and subjecting ourselves to a 'black box'
>> closed-source
>> solution. However, we've recently tried out the Huygens deconvolution
>> software, and the results seem quite promising, possibly an
>> improvement over other methods we've tried. I like good images, but I
>> don't like black boxes, and I like to understand my data processing.
>>
>> 1. Is the exact algorithm used in Huygens transparently documented
>> anywhere? I spent a few hours searching today, but if it's out there,
>> I missed it.
>>
>> 2. Is there a clear winner for deconvolution algorithms? What should
>> I be using?
>>
>> 3. Are there other deconvolution software packages I should consider?
>> Ideally I'm looking for software based on clearly-documented algorithms.
>>
>> Thanks for the help.
>>
>> -Andrew York
>> NIH/NIBIB
>>
>

--
Managing Director

Huygens SVI

tel: +31 (0)35 642 16 26

fax:  +31 (0)35 683 79 71

skype: gittahamel

cell: +31(0)618 021272

Visiting address

Laapersveld 63,
1213 VB Hilversum,
The Netherlands
Kevin Ryan Kevin Ryan
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Deconvolution advice - Commercial response

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

**COMMERCIAL RESPONSE, MEDIA CYBERNETICS - AUTOQUANT **

Jay is correct that imaging modes such as SIM will not have PSFs matching ordinary microscopic modes. For deconvolving novel imaging methods such as SIM, the first and best approach is probably to use a fixed, measured PSF:

* Image a point object under the same conditions as your desired sample (and that includes your _embedding media_, as the refractive index of your media is a critical element of image formation!), then use that as a model for the system PSF. That should incorporate all elements of the imaging system in the transform between the point object and the observed PSF. The better the PSF image, the better final reconstruction will be.

As an example, AutoQuant has been used in this fashion for computed holographic microscopy (Rosen and Brooker 2008, "Non-scanning motionless fluorescence three-dimensional holographic microscopy", http://citeseerx.ist.psu.edu/viewdoc/download?doi=10.1.1.153.3691&rep=rep1&type=pdf), with quite good results. The PSF for that system is notably _not_ a standard widefield or confocal blur.


* Blind deconvolution, where the PSF is optimized along with the data, _can_ be done with novel microscopic imaging modes by starting with a measured PSF and disabling PSF constraints (which are the constraints appropriate for widefield/confocal data - radial symmetry, energy conservation, etc). That would, however, be an experimental approach, and should be checked for ground truth against known specimens to see if it is appropriate for your data. Examine both the resulting deconvolution for known structures (did it remove blur, improve resolution?), and the resulting PSF for comparison to using SIM on a point object (does the final PSF have the general form of a measured point object?). This may provide a reconstruction improvement over a measured PSF by incorporating information from the observed sample.  


Kevin Ryan
Media Cybernetics, Inc.
 


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Unruh, Jay
Sent: Tuesday, October 02, 2012 10:18 AM
To: [hidden email]
Subject: Re: Deconvolution advice

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi all,

Deconvolution for SIM is a very different story than for other techniques.  SIM by default uses an inverse filter in its reconstruction to recombine the shifted components of the fourier transform without enhancing the high frequency noise.  Typical SIM software has a noise parameter for the wiener filter.  If you set this filter low, you start to see a rippling pattern in the noise and eventually in the high signal regions.  As far as I can tell, no one has tried to use more advanced (poisson noise driven) algorithms for this problem.  

If one assumes that Autoquant is using some variant of the algorithm shown in Tim Holmes' Handbook of Biological Confocal  chapter, then this algorithm is not immediately applicable to the SIM problem.  The algorithm update function involves dividing the original image by the convolved object guess and then convolving that ratio with the reflected psf and finally multiplying by the object guess.  Given that the raw SIM image is actually a frequency modulated image in a particular direction, it is not clear how the original ratio would be generated.  Richardson-Lucy has a similar problem.  Would you deconvolve before reconstructing?  This is the only circumstance under which the noise could be considered poisson.  In that case, can the original PSF be used?  I'm not entirely certain that the deconvolution would preserve the frequency modulation in the image.  In addition, the original reconstruction algorithm would require a second step of deconvolution during the reconstruction step--not sure how the noise parameter should be chosen after initial deconvolution.

Jay

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Gitta Hamel
Sent: Monday, October 01, 2012 8:42 AM
To: [hidden email]
Subject: Re: Deconvolution advice

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

**commercial response**


Hello Andrew,

It's fully understandable that people want to know the scientific grounds when using Huygens.

For the full list of articles I refer to http://www.svi.nl/HuygensReferences at which the relevant papers are at the bottom of the page and mostly written during the years 1996-1998.
There are much more articles that ought to be included so your question shows that we must give more attention to this topic.

With best wishes,

Gitta Hamel

****************************************
Gitta Hamel
Managing Director Scientific Volume Imaging bv Developers of the *HUYGENS* software The Netherlands
phone: ++ 31 35 6 42 16 26
*****************************************


^SVI Customer support: mail us your questions [hidden email] <mailto:[hidden email]>or find answers online in our Huygens WIKI:www.svi.nl/FrontPage <http://%20www.svi.nl/FrontPage>



On 09/29/2012 04:00 PM, Gens, John Scott wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Andrew-
>
> You might want to get in touch with Jim McNally.  Last I heard he was
> at NIH-NCI.
>
> Some of his older papers on deconvoltion algorithms are below, but he
> can probably point you towards more recent information.
>
> http://www.ncbi.nlm.nih.gov/pubmed/10579932
>
> http://www.ncbi.nlm.nih.gov/pubmed/11541650   ( in particular, fig.2  
> compared  a 3D image processed by three different algorithms)
>
>
> Quoting Andrew York <[hidden email]>:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hello, I'm looking for advice and information about deconvolution,
>> especially from those with first-hand experience.
>>
>> Traditionally, one of the processing steps in structured illumination
>> microscopy is deconvolution. For our SIM, we decided to use an
>> open-source
>> solution:
>> https://sites.google.com/site/piotrwendykier/software/deconvolution/p
>> aralleliterativedeconvolution
>>
>>
>> This seemed like a nice tradeoff between reinventing the wheel with
>> our own deconvolution code, and subjecting ourselves to a 'black box'
>> closed-source
>> solution. However, we've recently tried out the Huygens deconvolution
>> software, and the results seem quite promising, possibly an
>> improvement over other methods we've tried. I like good images, but I
>> don't like black boxes, and I like to understand my data processing.
>>
>> 1. Is the exact algorithm used in Huygens transparently documented
>> anywhere? I spent a few hours searching today, but if it's out there,
>> I missed it.
>>
>> 2. Is there a clear winner for deconvolution algorithms? What should
>> I be using?
>>
>> 3. Are there other deconvolution software packages I should consider?
>> Ideally I'm looking for software based on clearly-documented algorithms.
>>
>> Thanks for the help.
>>
>> -Andrew York
>> NIH/NIBIB
>>
>

--
Managing Director

Huygens SVI

tel: +31 (0)35 642 16 26

fax:  +31 (0)35 683 79 71

skype: gittahamel

cell: +31(0)618 021272

Visiting address

Laapersveld 63,
1213 VB Hilversum,
The Netherlands
######################################################################################
CONFIDENTIALITY NOTICE:
This email transmission and its attachments contain confidential and proprietary information
of Princeton Instruments, Acton Research, Media Cybernetics and their affiliates and is
intended for the exclusive and confidential use of the intended recipient. Any use, dissemination,
printing, or copying of this transmission and its attachment(s) is strictly prohibited. If you
are not the intended recipient, please do not read, print, copy, distribute or take action in
reliance upon this message.  If you have received this in error, please notify the sender immediately
by telephone or return email and promptly delete all copies of the original transmission and its
attachments from your computer system.
#######################################################################################
Lutz Schaefer Lutz Schaefer
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Deconvolution advice

In reply to this post by Unruh, Jay
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hello all,

Jay is correct, a regularized inverse filter of the separated (shifted)
orders can be used followed by magnitude (and/or linear) reconstruction. To
the best of my knowledge, this principle is implemented in the Zeiss ZEN and
AxioVision software. It is also possible to use a Gauss likelihood iterative
algorithm, but to no advantage as the extracted orders cannot be constrained
to positivity.
A realization of Poisson likelihood will, as Jay says proof difficult, owing
to its inherent nonlinearity.

Best Regards
Lutz

__________________________________
L u t z   S c h a e f e r
Sen. Scientist
Mathematical modeling / Image processing
Advanced Imaging Methodology Consultation
16-715 Doon Village Rd.
Kitchener, ON, N2P 2A2, Canada
Phone/Fax: +1 519 894 8870
Email:     [hidden email]
___________________________________

--------------------------------------------------
From: "Unruh, Jay" <[hidden email]>
Sent: Tuesday, October 02, 2012 10:17
To: <[hidden email]>
Subject: Re: Deconvolution advice

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi all,
>
> Deconvolution for SIM is a very different story than for other techniques.
> SIM by default uses an inverse filter in its reconstruction to recombine
> the shifted components of the fourier transform without enhancing the high
> frequency noise.  Typical SIM software has a noise parameter for the
> wiener filter.  If you set this filter low, you start to see a rippling
> pattern in the noise and eventually in the high signal regions.  As far as
> I can tell, no one has tried to use more advanced (poisson noise driven)
> algorithms for this problem.
>
> If one assumes that Autoquant is using some variant of the algorithm shown
> in Tim Holmes' Handbook of Biological Confocal  chapter, then this
> algorithm is not immediately applicable to the SIM problem.  The algorithm
> update function involves dividing the original image by the convolved
> object guess and then convolving that ratio with the reflected psf and
> finally multiplying by the object guess.  Given that the raw SIM image is
> actually a frequency modulated image in a particular direction, it is not
> clear how the original ratio would be generated.  Richardson-Lucy has a
> similar problem.  Would you deconvolve before reconstructing?  This is the
> only circumstance under which the noise could be considered poisson.  In
> that case, can the original PSF be used?  I'm not entirely certain that
> the deconvolution would preserve the frequency modulation in the image.
> In addition, the original reconstruction algorithm would require a second
> step of deconvolution during the reconstruction step--not sure how the
> noise parameter should be chosen after initial deconvolution.
>
> Jay
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]]
> On Behalf Of Gitta Hamel
> Sent: Monday, October 01, 2012 8:42 AM
> To: [hidden email]
> Subject: Re: Deconvolution advice
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> **commercial response**
>
>
> Hello Andrew,
>
> It's fully understandable that people want to know the scientific grounds
> when using Huygens.
>
> For the full list of articles I refer to
> http://www.svi.nl/HuygensReferences at which the relevant papers are at
> the bottom of the page and mostly written during the years 1996-1998.
> There are much more articles that ought to be included so your question
> shows that we must give more attention to this topic.
>
> With best wishes,
>
> Gitta Hamel
>
> ****************************************
> Gitta Hamel
> Managing Director Scientific Volume Imaging bv Developers of the *HUYGENS*
> software The Netherlands
> phone: ++ 31 35 6 42 16 26
> *****************************************
>
>
> ^SVI Customer support: mail us your questions [hidden email]
> <mailto:[hidden email]>or find answers online in our Huygens
> WIKI:www.svi.nl/FrontPage <http://%20www.svi.nl/FrontPage>
>
>
>
> On 09/29/2012 04:00 PM, Gens, John Scott wrote:
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Andrew-
>>
>> You might want to get in touch with Jim McNally.  Last I heard he was
>> at NIH-NCI.
>>
>> Some of his older papers on deconvoltion algorithms are below, but he
>> can probably point you towards more recent information.
>>
>> http://www.ncbi.nlm.nih.gov/pubmed/10579932
>>
>> http://www.ncbi.nlm.nih.gov/pubmed/11541650   ( in particular, fig.2
>> compared  a 3D image processed by three different algorithms)
>>
>>
>> Quoting Andrew York <[hidden email]>:
>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> Hello, I'm looking for advice and information about deconvolution,
>>> especially from those with first-hand experience.
>>>
>>> Traditionally, one of the processing steps in structured illumination
>>> microscopy is deconvolution. For our SIM, we decided to use an
>>> open-source
>>> solution:
>>> https://sites.google.com/site/piotrwendykier/software/deconvolution/p
>>> aralleliterativedeconvolution
>>>
>>>
>>> This seemed like a nice tradeoff between reinventing the wheel with
>>> our own deconvolution code, and subjecting ourselves to a 'black box'
>>> closed-source
>>> solution. However, we've recently tried out the Huygens deconvolution
>>> software, and the results seem quite promising, possibly an
>>> improvement over other methods we've tried. I like good images, but I
>>> don't like black boxes, and I like to understand my data processing.
>>>
>>> 1. Is the exact algorithm used in Huygens transparently documented
>>> anywhere? I spent a few hours searching today, but if it's out there,
>>> I missed it.
>>>
>>> 2. Is there a clear winner for deconvolution algorithms? What should
>>> I be using?
>>>
>>> 3. Are there other deconvolution software packages I should consider?
>>> Ideally I'm looking for software based on clearly-documented algorithms.
>>>
>>> Thanks for the help.
>>>
>>> -Andrew York
>>> NIH/NIBIB
>>>
>>
>
> --
> Managing Director
>
> Huygens SVI
>
> tel: +31 (0)35 642 16 26
>
> fax:  +31 (0)35 683 79 71
>
> skype: gittahamel
>
> cell: +31(0)618 021272
>
> Visiting address
>
> Laapersveld 63,
> 1213 VB Hilversum,
> The Netherlands
George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Deconvolution advice ... MSIM is "a lot like a point-scanning confocal with twice the resolution and higher signal-to-noise"

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Jay and Lutz,

I believe Andrew is not going to deconvolve the stripes, instead he is
going to deconvolve the points - see video 1 at

http://code.google.com/p/msim/

the ... in the subject line is from:

    That's what 'msim' does.

    Specifically, we built a new kind of fluorescence microscope, which
    we call a 'multifocal structured illumination microscope' (MSIM).
    Distinguishing features:

        * 3D superresolution (145 nm transverse, 400 nm axial resolution)
        * Uses standard fluorescent probes (like GFP), currently
          configured for 488 and 561 nm illumination
        * Works in thick samples (>50 microns). Previous SIM scopes
          couldn't do this.
        * Reasonable speed (1 2D slice per second)
        * Live-cell compatible (tens or hundreds of 3D volumes)
        * Low photobleaching (similar to a spinning disk confocal)
        * 50x50 micron field of view

    Functionally, it's a lot like a point-scanning confocal with twice
    the resolution and higher signal-to-noise, at the cost of digital
    post-processing.


His Nature Methods paper
http://www.nature.com/nmeth/journal/v9/n7/full/nmeth.2025.html  mentions
that the 3D superresolution values he measured, 145 nm XY, 300 nm Z, was
with a standard objective lens - not hand picked:

    "The apparent full-width at half maximum (FWHM) intensity of
    microtubules in MSIM images was 145 nm, a twofold improvement
    compared to wide-field imaging (Fig. 1d). Similar experiments on
    110-nm beads confirmed this result (MSIM FWHM of 146 nm ± 15 nm
    versus wide-field FWHM of 284 nm ± 32 nm (± s.d.), n = 80 beads,
    Supplementary Fig. 7). The acquisition time for the 48 ?m × 49 ?m
    field (Fig. 1a) was ~1 s, more than a 6,500-fold improvement over
    ISM, assuming the same 222-Hz raw frame rate."

and,

    Like existing 3D SIM16, MSIM provides resolution doubling by
    increasing the highest spatial frequencies encoded in the raw data,
    followed by deconvolution. However, the absolute resolutions we
    report are slightly lower than existing 3D SIM, which we attribute
    to two factors. First, achieving the highest possible resolution
    relies on obtaining diffraction-limited wide-field performance
    before (M)SIM is applied. Although we attempted this, the wide-field
    point spread functions (PSFs) we measured (Supplementary Fig. 7) had
    FWHM (~280 nm) larger than theoretical predictions. It is customary
    in SIM to screen many objectives and use only the one with highest
    resolution. We did not do this. Furthermore, we suspect that
    aberration caused by warping of the dichroic in our microscope was
    the major cause of departure from ideal performance, as we noticed
    that small adjustments of the mounting screws that held the dichroic
    drastically altered the PSF. A thicker, stiffer dichroic may
    mitigate this issue. Second, the excitation patterns in previous SIM
    implementations concentrate energy at the very highest lateral
    spatial frequencies in the objective pass-band. In contrast, the
    multifocal excitation patterns we used contain all spatial
    frequencies permitted by the objective. We might thus expect
    line-excitation--based SIM to provide a greater SNR at high spatial
    frequencies than MSIM, enabling higher resolution. We note the
    point-like nature of our excitation pattern offers a route to even
    higher resolution, perhaps by exploiting the nonlinearities provided
    by reversible, switchable fluorescent probes29.
    Compared to previous SIM implementations, our system is simpler to
    build. We add only a DMD, a telescope and a fast camera to a
    conventional wide-field microscope. Our illumination path has no
    moving parts and is insensitive to polarization. Finally, we do not
    modify the illumination path for multicolor operation, unlike
    existing SIM implementations30.


Upshot: MSIM is likely to get even  better!

George


On 10/2/2012 12:40 PM, Lutz Schaefer wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hello all,
>
> Jay is correct, a regularized inverse filter of the separated
> (shifted) orders can be used followed by magnitude (and/or linear)
> reconstruction. To the best of my knowledge, this principle is
> implemented in the Zeiss ZEN and AxioVision software. It is also
> possible to use a Gauss likelihood iterative algorithm, but to no
> advantage as the extracted orders cannot be constrained to positivity.
> A realization of Poisson likelihood will, as Jay says proof difficult,
> owing to its inherent nonlinearity.
>
> Best Regards
> Lutz
>
> __________________________________
> L u t z   S c h a e f e r
> Sen. Scientist
> Mathematical modeling / Image processing
> Advanced Imaging Methodology Consultation
> 16-715 Doon Village Rd.
> Kitchener, ON, N2P 2A2, Canada
> Phone/Fax: +1 519 894 8870
> Email:     [hidden email]
> ___________________________________
>
> --------------------------------------------------
> From: "Unruh, Jay" <[hidden email]>
> Sent: Tuesday, October 02, 2012 10:17
> To: <[hidden email]>
> Subject: Re: Deconvolution advice
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hi all,
>>
>> Deconvolution for SIM is a very different story than for other
>> techniques. SIM by default uses an inverse filter in its
>> reconstruction to recombine the shifted components of the fourier
>> transform without enhancing the high frequency noise.  Typical SIM
>> software has a noise parameter for the wiener filter.  If you set
>> this filter low, you start to see a rippling pattern in the noise and
>> eventually in the high signal regions.  As far as I can tell, no one
>> has tried to use more advanced (poisson noise driven) algorithms for
>> this problem.
>>
>> If one assumes that Autoquant is using some variant of the algorithm
>> shown in Tim Holmes' Handbook of Biological Confocal  chapter, then
>> this algorithm is not immediately applicable to the SIM problem.  The
>> algorithm update function involves dividing the original image by the
>> convolved object guess and then convolving that ratio with the
>> reflected psf and finally multiplying by the object guess.  Given
>> that the raw SIM image is actually a frequency modulated image in a
>> particular direction, it is not clear how the original ratio would be
>> generated.  Richardson-Lucy has a similar problem.  Would you
>> deconvolve before reconstructing?  This is the only circumstance
>> under which the noise could be considered poisson.  In that case, can
>> the original PSF be used?  I'm not entirely certain that the
>> deconvolution would preserve the frequency modulation in the image.
>> In addition, the original reconstruction algorithm would require a
>> second step of deconvolution during the reconstruction step--not sure
>> how the noise parameter should be chosen after initial deconvolution.
>>
>> Jay
>>
>> -----Original Message-----
>> From: Confocal Microscopy List
>> [mailto:[hidden email]] On Behalf Of Gitta Hamel
>> Sent: Monday, October 01, 2012 8:42 AM
>> To: [hidden email]
>> Subject: Re: Deconvolution advice
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> **commercial response**
>>
>>
>> Hello Andrew,
>>
>> It's fully understandable that people want to know the scientific
>> grounds when using Huygens.
>>
>> For the full list of articles I refer to
>> http://www.svi.nl/HuygensReferences at which the relevant papers are
>> at the bottom of the page and mostly written during the years 1996-1998.
>> There are much more articles that ought to be included so your
>> question shows that we must give more attention to this topic.
>>
>> With best wishes,
>>
>> Gitta Hamel
>>
>> ****************************************
>> Gitta Hamel
>> Managing Director Scientific Volume Imaging bv Developers of the
>> *HUYGENS* software The Netherlands
>> phone: ++ 31 35 6 42 16 26
>> *****************************************
>>
>>
>> ^SVI Customer support: mail us your questions [hidden email]
>> <mailto:[hidden email]>or find answers online in our Huygens
>> WIKI:www.svi.nl/FrontPage <http://%20www.svi.nl/FrontPage>
>>
>>
>>
>> On 09/29/2012 04:00 PM, Gens, John Scott wrote:
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> Andrew-
>>>
>>> You might want to get in touch with Jim McNally.  Last I heard he was
>>> at NIH-NCI.
>>>
>>> Some of his older papers on deconvoltion algorithms are below, but he
>>> can probably point you towards more recent information.
>>>
>>> http://www.ncbi.nlm.nih.gov/pubmed/10579932
>>>
>>> http://www.ncbi.nlm.nih.gov/pubmed/11541650   ( in particular, fig.2
>>> compared  a 3D image processed by three different algorithms)
>>>
>>>
>>> Quoting Andrew York <[hidden email]>:
>>>
>>>> *****
>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>> *****
>>>>
>>>> Hello, I'm looking for advice and information about deconvolution,
>>>> especially from those with first-hand experience.
>>>>
>>>> Traditionally, one of the processing steps in structured illumination
>>>> microscopy is deconvolution. For our SIM, we decided to use an
>>>> open-source
>>>> solution:
>>>> https://sites.google.com/site/piotrwendykier/software/deconvolution/p
>>>> aralleliterativedeconvolution
>>>>
>>>>
>>>> This seemed like a nice tradeoff between reinventing the wheel with
>>>> our own deconvolution code, and subjecting ourselves to a 'black box'
>>>> closed-source
>>>> solution. However, we've recently tried out the Huygens deconvolution
>>>> software, and the results seem quite promising, possibly an
>>>> improvement over other methods we've tried. I like good images, but I
>>>> don't like black boxes, and I like to understand my data processing.
>>>>
>>>> 1. Is the exact algorithm used in Huygens transparently documented
>>>> anywhere? I spent a few hours searching today, but if it's out there,
>>>> I missed it.
>>>>
>>>> 2. Is there a clear winner for deconvolution algorithms? What should
>>>> I be using?
>>>>
>>>> 3. Are there other deconvolution software packages I should consider?
>>>> Ideally I'm looking for software based on clearly-documented
>>>> algorithms.
>>>>
>>>> Thanks for the help.
>>>>
>>>> -Andrew York
>>>> NIH/NIBIB
>>>>
>>>
>>
>> --
>> Managing Director
>>
>> Huygens SVI
>>
>> tel: +31 (0)35 642 16 26
>>
>> fax:  +31 (0)35 683 79 71
>>
>> skype: gittahamel
>>
>> cell: +31(0)618 021272
>>
>> Visiting address
>>
>> Laapersveld 63,
>> 1213 VB Hilversum,
>> The Netherlands
>
Brian Northan Brian Northan
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Deconvolution advice

In reply to this post by Lutz Schaefer
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Andrew

As previously discussed there are a multitude of issues to consider
when deconvolving the images from your customized microscope such as
choosing an algorithm appropriate for the noise model.  While
investigating that you would also want to understand any noise related
user-adjustable parameters the algorithms may have.

After you've worked through the noise handling of the algorithm you
would want to look very carefully at the PSF generation.  I was with
AutoQuant/MediaCy for many years and worked on (among other things)
theoretical psf and aberration models, so I have some thoughts on
these issues.

As others have mentioned for a novel modality, if no software is
available for calculating a theoretical PSF, you would have to use a
measured PSF.   From being involved in deconvolution in the past and
attending the same workshops as SVI I am moderately familiar with
Huygens.  One possibility is that the differences you are seeing
between the Huygens implementation and the Parallel Iterative open
source implementation could be partially due to PSF processing.  You
should ask SVI more about this part.

As you've probably experienced the measured PSF needs to be
preconditioned.   At the very least it should be centred and made the
same size as the image.  It is sometimes denoised and made (at least
laterally-radially) symmetric.  There are many variations of psf
preprocessing routines.  Some involve preconditioning a single bead.
Some average multiple beads.  Another alternative is to solve for the
PSF using a "reverse deconvolution" on an object of known size and
shape (usually a field of beads which in this case do not need to be
sub resolution).

Huygens has the option of solving for the PSF through a reverse
deconvolution scheme they call "distilling".

So it could be that the PSF pre-processing and/or distilling (if you
are using it) in the Huygens software is making a difference.  I don't
know for sure but it would be something to investigate.  You should
see if Huygens allows you to export the preprocessed/distilled PSF and
if so take a look at it.

Brian