Hello everybody,
I try to measure FRET with a FLIM system from Becker and Hickl containing a MCP detector from Hamamatsu (R3809U-50). This detector is very temperature sensitive (increase of dark counts with increasing temperature). Our room is air conditioned so the room temperature is constant.
When I measure CFP over a couple of houres (I measure live cells and allways take a fresh probe) I observe a decrease in lifetime (0.4 ns in 5 hours).
How can I avoid this problem??
Has anybody else observed this?
I think it could be the detector and that it is getting hot over the time. Could I cool it down somehow?
Thanks for any helpful advise.
Anke
|
|
|
Hello Anke:
Try the Hamamatsu C4878 cooler (you will need th holder as well). It keeps the MCP-PMT at -30 celsius. Read the intructions carefully on operating the cooler. No financial interest. Regards, Sophie ____________________________________________________ Sophie M. K. Brunet, Ph. D. Research Officer Optical Spectroscopy, Laser Systems and Applications [hidden email] 306-966-1719 (office) 306-966-1702 (fax) ____________________________________________________ Saskatchewan Structural Sciences Centre University of Saskatchewan Thorvaldson Bldg. 110 Science Place Saskatoon, Sk S7N 5C9 ____________________________________________________ Quoting Anke Bielicke <[hidden email]>: > Hello everybody, > > I try to measure FRET with a FLIM system from Becker and Hickl containing a > MCP detector from Hamamatsu (R3809U-50). This detector is very temperature > sensitive (increase of dark counts with increasing temperature). Our room is > air conditioned so the room temperature is constant. > > When I measure CFP over a couple of houres (I measure live cells and allways > take a fresh probe) I observe a decrease in lifetime (0.4 ns in 5 hours). > > How can I avoid this problem?? > Has anybody else observed this? > I think it could be the detector and that it is getting hot over the time. > Could I cool it down somehow? > > > Thanks for any helpful advise. > > Anke > > |
|
Michelle Digman |
|
|
In reply to this post by Anke Bielicke
Hi Anke,
I am copying an abstract from a paper describing the photosenstivity of CFP . Perhaps the problem is not the detector but the CFP itself. I have also seen this decrease in lifetime with CFP. If you have Ceruleun it is not as photosensitive as CFP. Best, Michelle Research Article Sensitivity of CFP/YFP and GFP/mCherry pairs to donor photobleaching on FRET determination by fluorescence lifetime imaging microscopy in living cells Marc Tramier, Morad Zahid, Jean-Claude Mevel, Marie-Jo Masse, Maïté Coppey-Moisan * Abstract Fluorescent protein-based FRET is a powerful method for visualizing protein-protein interactions and biochemical reactions in living cells. It can be difficult, however, to avoid photobleaching when observing fluorescent cells under the microscope, especially those expressing CFP. We compared the sensitivity of two protein-based FRET pairs to light-induced fluorescence changes in the donor, on FRET determination by fluorescence lifetime imaging microscopy (FLIM). Thanks to the very low excitation light levels of the time- and space-correlated single photon counting (TSCSPC) method, FLIM acquisitions were achieved without donor photobleaching. Here, we show that photobleaching of CFP by a mercury lamp under the microscope induced a decrease in the mean fluorescence lifetime, which interfered with FRET determination between CFP and YFP. Importantly, the range of light-induced variation of the mean fluorescence lifetime of CFP was not proportional to the decrease in the steady state fluorescence intensity and varied from cell to cell. The choice of the CFP/YFP pair therefore requires that the cells be observed and analyzed at very low light levels during the whole FRET experiment. In contrast, the GFP/mCherry pair provided an accurate FRET measurement by FLIM, even if some GFP photobleaching took place. We thus demonstrate that CFP can be an unreliable donor for FRET determination in living cells, due to its photosensitivity properties. We demonstrate that the GFP/mCherry pair is better suited for FRET measurement by FLIM in living cells than the CFP/YFP pair. Microsc. Res. Tech., 2006. © 2006 Wiley-Liss, Inc. -- Michelle Digman, Ph.D. Director, Optical Biology Core University of California, Irvine McGaugh Hall 4443 (949) 824-3856 (OBC office) (949) 282-8220 (mobile) http://dbc.bio.uci.edu/index.html Laboratory for Fluorescence Dynamics |
|
Stroetz, Randolph W., R.R.T., R.C.P. |
|
|
In reply to this post by S. Brunet
Has anyone seen or used an adapter allowing an infinity-corrected microscope objective to be mounted on a digital SLR camera? Thanks, Randy |
|
|
Leica objective threads are pretty darn close to a c-mount
specification; from there you can usually get other adaptors (i.e. bayonet-to c-mount, etc). Not sure what the other manufacturers use. ----- Original Message ----- From: "Stroetz, Randolph W., R.R.T., R.C.P." <[hidden email]> Date: Tuesday, July 28, 2009 4:25 pm Subject: Odd camera adapter To: [hidden email] > > Has anyone seen or used an adapter allowing an infinity-corrected > microscope objective to be mounted on a digital SLR camera? > > Thanks, Randy > |
|
|
An adapter from one of the optical companies like Edmund Optics will
probably do this, you'll also need a tube lens in the adapter - check the microscopy sections of these catalogues. We've been looking into this ourselves, and may go the route of purchasing one of their very long working distance objectives with adapter to do time-lapse imaging of root growth. cheers, Rosemary Rosemary White CSIRO Plant Industry GPO Box 1600 Canberra, ACT 2601 Australia ph 61 2 6246 5475 fx 61 2 6246 5334 On 29/07/09 6:31 AM, "Andy Resnick" <[hidden email]> wrote: > Leica objective threads are pretty darn close to a c-mount > specification; from there you can usually get other adaptors (i.e. > bayonet-to c-mount, etc). Not sure what the other manufacturers use. > > ----- Original Message ----- > From: "Stroetz, Randolph W., R.R.T., R.C.P." <[hidden email]> > Date: Tuesday, July 28, 2009 4:25 pm > Subject: Odd camera adapter > To: [hidden email] > >> >> Has anyone seen or used an adapter allowing an infinity-corrected >> microscope objective to be mounted on a digital SLR camera? >> >> Thanks, Randy >> |
|
|
As Rosemary suggested, the Edmund Optics catalog has c-mount tubes with
an adapter for the objective mount, and a c-mount to t-mount that will then attach to a Nikon or Canon mount. In my own experience with a non-infinity lens at 140mm from back of lens to camera chip, everything worked well. For infinity objectives, you might be able to get away with a 175 or 200mm focal length meniscus for the tube lens. With Edmund Optics c-mount tubes, you may have to get adapters to include the lens in the path (the lenses may not fit in the tube at the diameters sold). Jerry Rosemary White wrote: > An adapter from one of the optical companies like Edmund Optics will > probably do this, you'll also need a tube lens in the adapter - check the > microscopy sections of these catalogues. We've been looking into this > ourselves, and may go the route of purchasing one of their very long working > distance objectives with adapter to do time-lapse imaging of root growth. > cheers, > Rosemary > > Rosemary White > CSIRO Plant Industry > GPO Box 1600 > Canberra, ACT 2601 > Australia > > ph 61 2 6246 5475 > fx 61 2 6246 5334 > > > > On 29/07/09 6:31 AM, "Andy Resnick" <[hidden email]> wrote: > > >> Leica objective threads are pretty darn close to a c-mount >> specification; from there you can usually get other adaptors (i.e. >> bayonet-to c-mount, etc). Not sure what the other manufacturers use. >> >> ----- Original Message ----- >> From: "Stroetz, Randolph W., R.R.T., R.C.P." <[hidden email]> >> Date: Tuesday, July 28, 2009 4:25 pm >> Subject: Odd camera adapter >> To: [hidden email] >> >> >>> Has anyone seen or used an adapter allowing an infinity-corrected >>> microscope objective to be mounted on a digital SLR camera? >>> >>> Thanks, Randy >>> >>> > > > -- Jerry (Gerald) Sedgewick Core Facility Director, Biomedical Image Processing Lab (BIPL) University of Minnesota, Department of Neuroscience 1-205 Hasselmo Hall 312 Church St. S.E. Minneapolis, MN 55455 612-624-6607 [hidden email] http://www.bipl.umn.edu Author: "Scientific Imaging with Photoshop: Methods, Measurement and Output." Rawlight.com (dba "Sedgewick Initiatives") 965 Cromwell Avenue Saint Paul, MN 55114 651-788-2261 [hidden email] http://www.rawlight.com http://www.jerrysedgewick.com --- Get FREE High Speed Internet from USFamily.Net! -- http://www.usfamily.net/mkt-freepromo.html --- |
|
|
In reply to this post by Stroetz, Randolph W., R.R.T., R.C.P.
You might want to check http://www.scopetronics.com I use their MAXVIEW
PLUS with many digital cameras. |
|
Edelmann, Richard E. Dr. |
|
|
In reply to this post by Stroetz, Randolph W., R.R.T., R.C.P.
Once upon a time Canon used to make one for their 35mm cameras (1970's-1980's). The lens mounts for the dSLR are STANDARD lens camera manufacturer's mounts (i.e. F-mounts, Canon Mount, etc.). I have tested a number of old (pre-digital) lenses on the Nikon dSLR's and they work well (but no auto-focus obviously). So you could look for an old Canon microscope objective mount.
________________________________________ From: Confocal Microscopy List [[hidden email]] On Behalf Of Stroetz, Randolph W., R.R.T., R.C.P. [[hidden email]] Sent: Tuesday, July 28, 2009 4:25 PM To: [hidden email] Subject: Odd camera adapter Has anyone seen or used an adapter allowing an infinity-corrected microscope objective to be mounted on a digital SLR camera? Thanks, Randy |
|
Edelmann, Richard E. Dr. |
|
|
In reply to this post by Stroetz, Randolph W., R.R.T., R.C.P.
Quick searching:
http://www.alanwood.net/photography/olympus/objective-lens-mount.html (Oh, and yes the older Canon made mounts were designed to be used with a macro-bellow system for proper focusing). ________________________________________ From: Confocal Microscopy List [[hidden email]] On Behalf Of Stroetz, Randolph W., R.R.T., R.C.P. [[hidden email]] Sent: Tuesday, July 28, 2009 4:25 PM To: [hidden email] Subject: Odd camera adapter Has anyone seen or used an adapter allowing an infinity-corrected microscope objective to be mounted on a digital SLR camera? Thanks, Randy |
|
|
Diagnostic Instruments also makes adapters for "unique" applications.
Not quite sure how costs would compare to Edmond's for example. http://www.diaginc.com/accessories/coupler/ No commercial interest. Brian D Armstrong PhD Light Microscopy Core Manager Beckman Research Institute City of Hope Dept of Neuroscience 1450 E Duarte Rd Duarte, CA 91010 626-256-4673 x62872 http://www.cityofhope.org/research/support/Light-Microscopy-Digital-Imag ing/Pages/default.aspx -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Edelmann, Richard E. Dr. Sent: Wednesday, July 29, 2009 8:55 AM To: [hidden email] Subject: Re: Odd camera adapter Quick searching: http://www.alanwood.net/photography/olympus/objective-lens-mount.html (Oh, and yes the older Canon made mounts were designed to be used with a macro-bellow system for proper focusing). ________________________________________ From: Confocal Microscopy List [[hidden email]] On Behalf Of Stroetz, Randolph W., R.R.T., R.C.P. [[hidden email]] Sent: Tuesday, July 28, 2009 4:25 PM To: [hidden email] Subject: Odd camera adapter Has anyone seen or used an adapter allowing an infinity-corrected microscope objective to be mounted on a digital SLR camera? Thanks, Randy --------------------------------------------------------------------- SECURITY/CONFIDENTIALITY WARNING: This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. --------------------------------------------------------------------- |
|
Michelle Digman |
|
|
In reply to this post by Michelle Digman
Dear List,
Announcing the 4th LFD Workshop in Advanced Fluorescence Imaging and Dynamics October 26-30, 2009 For full details on course content, registration procedures, and much more information, please see visit our website at http://www.lfd.uci.edu/workshop/ . Please note that there remain a limited number of laboratory training section spaces available. We assure our applicants that the computer training covers the exact same course material, and we also maintain a laboratory training waitlist. Thank you for forwarding this announcement to colleagues you think may be interested in our workshop! ***************************ATTENTION! AWARDS FOR STUDENTS!!************* Registration Awards for Students There will be 6 registration waivers for students. The winners will have the opportunity to present their work in a 10 minutes presentation during the workshop. Please submit a short abstract of your work (half page) and a support letter from your advisor before September 26, 2009, to: Enrico Gratton University of California, Irvine Biomedical Engineering Department 3210 Natural Sciences 2 Irvine, CA, 92697-2715 |