Dictyostelium confocal and heat

classic Classic list List threaded Threaded
♦
♦
10 messages Options
Oshel, Philip Eugene Oshel, Philip Eugene
Reply | Threaded
Open this post in threaded view
♦
♦
|

Dictyostelium confocal and heat

Confocalers,

Anyone here working on Dictyostelium? If yes, do you have problems
with heat from the illumination source killing the critters?
The cells are really sensitive to heat and light, so I'd like to hear
some actual experiences with different systems looking at live
Dictys, and how happy they were after you were done looking at them.
A regular fluorescence 'scope can easily kill them.

"Illumination source", as I'm wondering about both spinning-disc
systems that use Hg or Xe lamps and scanned-laser systems, point
scanners or line scanners. We're looking to get a new confocal, and
what kind of system (disc, point scanner, line scanner, millions of
nanognomes with LED flashlights, whatever) is an issue for the group
here.

I'm already dealing with vendors, so no vendor contacts off-list,
please. Technical information appropriate for the list is appreciated.

Thanks.

Phil
--
Philip Oshel
Microscopy Facility Supervisor
Biology Department
024C Brooks Hall
Central Michigan University
Mt. Pleasant, MI 48859
(989) 774-3576
Knecht, David Knecht, David
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Dictyostelium confocal and heat

Hi Philip- We have been wrestling with this problem for years.  We have heat reflection and heat absorption filters in the transmitted path of all our microscopes.  We always try to minimize light levels for both fluorescence and transmitted as that has a bigger effect than exposure time.  We have used various light sources and all are problematic.  Dicty is a great test of the efficiency of a confocal system, as their rounding up response is so easy to see.  We used that as a diagnostic in evaluating various systems.  In the end, they can all be made to work, but it comes down to how often you take images and how low you can turn down the light and still see your probe.  I think that in general, deconvolution or a laser scanning system in which you can open the pinhole and sacrifice some confocality are the best.  Spinning disk is good, but we just got one so I don't have as much experience yet.  Others have indicated it is an improvement and my initial tests indicate we can image faster.  Our Leica confocal is probably the worst in our stable because you cannot open the pinhole and decrease confocality, however we use it regularly for xyt imaging by just not collecting images as often.  Bottom line- no magic anywhere, only tradeoffs.   Dave

On Dec 12, 2008, at 8:37 AM, Philip Oshel wrote:

Confocalers,

Anyone here working on Dictyostelium? If yes, do you have problems with heat from the illumination source killing the critters?
The cells are really sensitive to heat and light, so I'd like to hear some actual experiences with different systems looking at live Dictys, and how happy they were after you were done looking at them. A regular fluorescence 'scope can easily kill them.

"Illumination source", as I'm wondering about both spinning-disc systems that use Hg or Xe lamps and scanned-laser systems, point scanners or line scanners. We're looking to get a new confocal, and what kind of system (disc, point scanner, line scanner, millions of nanognomes with LED flashlights, whatever) is an issue for the group here.

I'm already dealing with vendors, so no vendor contacts off-list, please. Technical information appropriate for the list is appreciated.

Thanks.

Phil
--
Philip Oshel
Microscopy Facility Supervisor
Biology Department
024C Brooks Hall
Central Michigan University
Mt. Pleasant, MI 48859
(989) 774-3576

Dr. David Knecht    
Department of Molecular and Cell Biology
Co-head Flow Cytometry and Confocal Microscopy Facility
U-3125
91 N. Eagleville Rd.
University of Connecticut
Storrs, CT 06269
860-486-2200
860-486-4331 (fax)


Phillips, Thomas E. Phillips, Thomas E.
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Dictyostelium confocal and heat

David – based on your experience with Dicty, what temperature do you think your confocals are getting the slides to?  Ideally your answer include the laser strength/scan speed/etc. thanks, Tom

 

 

 

Thomas E. Phillips, Ph.D

Professor of Biological Sciences

Chair, MU Faculty Council

Director, Molecular Cytology Core

2 Tucker Hall

University of Missouri

Columbia, MO 65211-7400

573-882-4712 (office)

573-882-0123 (fax)

[hidden email]

 

http://www.biology.missouri.edu/faculty/phillips.html
http://www.biotech.missouri.edu/mcc/

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of David Knecht
Sent: Friday, December 12, 2008 9:48 AM
To: [hidden email]
Subject: Re: Dictyostelium confocal and heat

 

Hi Philip- We have been wrestling with this problem for years.  We have heat reflection and heat absorption filters in the transmitted path of all our microscopes.  We always try to minimize light levels for both fluorescence and transmitted as that has a bigger effect than exposure time.  We have used various light sources and all are problematic.  Dicty is a great test of the efficiency of a confocal system, as their rounding up response is so easy to see.  We used that as a diagnostic in evaluating various systems.  In the end, they can all be made to work, but it comes down to how often you take images and how low you can turn down the light and still see your probe.  I think that in general, deconvolution or a laser scanning system in which you can open the pinhole and sacrifice some confocality are the best.  Spinning disk is good, but we just got one so I don't have as much experience yet.  Others have indicated it is an improvement and my initial tests indicate we can image faster.  Our Leica confocal is probably the worst in our stable because you cannot open the pinhole and decrease confocality, however we use it regularly for xyt imaging by just not collecting images as often.  Bottom line- no magic anywhere, only tradeoffs.   Dave

 

On Dec 12, 2008, at 8:37 AM, Philip Oshel wrote:



Confocalers,

Anyone here working on Dictyostelium? If yes, do you have problems with heat from the illumination source killing the critters?
The cells are really sensitive to heat and light, so I'd like to hear some actual experiences with different systems looking at live Dictys, and how happy they were after you were done looking at them. A regular fluorescence 'scope can easily kill them.

"Illumination source", as I'm wondering about both spinning-disc systems that use Hg or Xe lamps and scanned-laser systems, point scanners or line scanners. We're looking to get a new confocal, and what kind of system (disc, point scanner, line scanner, millions of nanognomes with LED flashlights, whatever) is an issue for the group here.

I'm already dealing with vendors, so no vendor contacts off-list, please. Technical information appropriate for the list is appreciated.

Thanks.

Phil
--
Philip Oshel
Microscopy Facility Supervisor
Biology Department
024C Brooks Hall
Central Michigan University
Mt. Pleasant, MI 48859
(989) 774-3576

 

Dr. David Knecht    

Department of Molecular and Cell Biology

Co-head Flow Cytometry and Confocal Microscopy Facility

U-3125

91 N. Eagleville Rd.

University of Connecticut

Storrs, CT 06269

860-486-2200

860-486-4331 (fax)



 

Michael Herron Michael Herron
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Dictyostelium confocal and heat

In reply to this post by Knecht, David
David,

Could you please keep us up-to-date with regards to your upcoming  
luck with the spinning disk system?  We are trying to make the same  
decision now.

Mike

On Dec 12, 2008, at 9:48 AM, David Knecht wrote:

> Hi Philip- We have been wrestling with this problem for years.  We  
> have heat reflection and heat absorption filters in the transmitted  
> path of all our microscopes.  We always try to minimize light  
> levels for both fluorescence and transmitted as that has a bigger  
> effect than exposure time.  We have used various light sources and  
> all are problematic.  Dicty is a great test of the efficiency of a  
> confocal system, as their rounding up response is so easy to see.  
> We used that as a diagnostic in evaluating various systems.  In the  
> end, they can all be made to work, but it comes down to how often  
> you take images and how low you can turn down the light and still  
> see your probe.  I think that in general, deconvolution or a laser  
> scanning system in which you can open the pinhole and sacrifice  
> some confocality are the best.  Spinning disk is good, but we just  
> got one so I don't have as much experience yet.  Others have  
> indicated it is an improvement and my initial tests indicate we can  
> image faster.  Our Leica confocal is probably the worst in our  
> stable because you cannot open the pinhole and decrease  
> confocality, however we use it regularly for xyt imaging by just  
> not collecting images as often.  Bottom line- no magic anywhere,  
> only tradeoffs.   Dave
>
> On Dec 12, 2008, at 8:37 AM, Philip Oshel wrote:
>
>> Confocalers,
>>
>> Anyone here working on Dictyostelium? If yes, do you have problems  
>> with heat from the illumination source killing the critters?
>> The cells are really sensitive to heat and light, so I'd like to  
>> hear some actual experiences with different systems looking at  
>> live Dictys, and how happy they were after you were done looking  
>> at them. A regular fluorescence 'scope can easily kill them.
>>
>> "Illumination source", as I'm wondering about both spinning-disc  
>> systems that use Hg or Xe lamps and scanned-laser systems, point  
>> scanners or line scanners. We're looking to get a new confocal,  
>> and what kind of system (disc, point scanner, line scanner,  
>> millions of nanognomes with LED flashlights, whatever) is an issue  
>> for the group here.
>>
>> I'm already dealing with vendors, so no vendor contacts off-list,  
>> please. Technical information appropriate for the list is  
>> appreciated.
>>
>> Thanks.
>>
>> Phil
>> --
>> Philip Oshel
>> Microscopy Facility Supervisor
>> Biology Department
>> 024C Brooks Hall
>> Central Michigan University
>> Mt. Pleasant, MI 48859
>> (989) 774-3576
>
> Dr. David Knecht
> Department of Molecular and Cell Biology
> Co-head Flow Cytometry and Confocal Microscopy Facility
> U-3125
> 91 N. Eagleville Rd.
> University of Connecticut
> Storrs, CT 06269
> 860-486-2200
> 860-486-4331 (fax)
>
>

Michael J. Herron,  U of MN, Dept. of Entomology
   [hidden email]
      612-624-3688 (office) 612-625-5299 (FAX)
Phillips, Thomas E. Phillips, Thomas E.
Reply | Threaded
Open this post in threaded view
♦
♦
|

Dictyostelium confocal and heat

In reply to this post by Oshel, Philip Eugene
David - based on your experience with Dicty, what temperature do you think your confocals are getting the slides to?  Ideally your answer include the laser strength/scan speed/etc. thanks, Tom



Thomas E. Phillips, Ph.D
Professor of Biological Sciences
Chair, MU Faculty Council
Director, Molecular Cytology Core
2 Tucker Hall
University of Missouri
Columbia, MO 65211-7400
573-882-4712 (office)
573-882-0123 (fax)
[hidden email]

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of David Knecht
Sent: Friday, December 12, 2008 9:48 AM
To: [hidden email]
Subject: Re: Dictyostelium confocal and heat

Hi Philip- We have been wrestling with this problem for years.  We have heat reflection and heat absorption filters in the transmitted path of all our microscopes.  We always try to minimize light levels for both fluorescence and transmitted as that has a bigger effect than exposure time.  We have used various light sources and all are problematic.  Dicty is a great test of the efficiency of a confocal system, as their rounding up response is so easy to see.  We used that as a diagnostic in evaluating various systems.  In the end, they can all be made to work, but it comes down to how often you take images and how low you can turn down the light and still see your probe.  I think that in general, deconvolution or a laser scanning system in which you can open the pinhole and sacrifice some confocality are the best.  Spinning disk is good, but we just got one so I don't have as much experience yet.  Others have indicated it is an improvement and my initial tests indicate we can image faster.  Our Leica confocal is probably the worst in our stable because you cannot open the pinhole and decrease confocality, however we use it regularly for xyt imaging by just not collecting images as often.  Bottom line- no magic anywhere, only tradeoffs.   Dave

On Dec 12, 2008, at 8:37 AM, Philip Oshel wrote:

Confocalers,

Anyone here working on Dictyostelium? If yes, do you have problems with heat from the illumination source killing the critters?
The cells are really sensitive to heat and light, so I'd like to hear some actual experiences with different systems looking at live Dictys, and how happy they were after you were done looking at them. A regular fluorescence 'scope can easily kill them.

"Illumination source", as I'm wondering about both spinning-disc systems that use Hg or Xe lamps and scanned-laser systems, point scanners or line scanners. We're looking to get a new confocal, and what kind of system (disc, point scanner, line scanner, millions of nanognomes with LED flashlights, whatever) is an issue for the group here.

I'm already dealing with vendors, so no vendor contacts off-list, please. Technical information appropriate for the list is appreciated.

Thanks.

Phil
--
Philip Oshel
Microscopy Facility Supervisor
Biology Department
024C Brooks Hall
Central Michigan University
Mt. Pleasant, MI 48859
(989) 774-3576

Dr. David Knecht    
Department of Molecular and Cell Biology
Co-head Flow Cytometry and Confocal Microscopy Facility
U-3125
91 N. Eagleville Rd.
University of Connecticut
Storrs, CT 06269
860-486-2200
860-486-4331 (fax)
Knecht, David Knecht, David
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Dictyostelium confocal and heat

I don't think this is about raising the temperature of the cells at all.  We would see a shift in focal plane as the coverslip shifts if the temp were increasing significantly.  I believe this is more about photodamage and photooxidation.  If the cells are in HL5 medium which is very fluorescent, the cells are very sensitive to fluorescent excitation.  If you change to LoFLo or buffer and allow them to recycle out their endosomes, they are dramatically less sensitive.  Developing cells that have been out of medium for 6-8 hours are even easier to image.  We also found that oxyrase will increase their resistance to light, indicating photooxidative damage.  I have a time lapse movie where the cells are imaged on the Leica, where they rapidly round up and then when you turn the laser down, they very rapidly begin to extend and move again.  We did some work with antioxidants, but not enough to find one that solves the problem.  Dave

On Dec 12, 2008, at 11:39 AM, Phillips, Thomas E. wrote:

David - based on your experience with Dicty, what temperature do you think your confocals are getting the slides to?  Ideally your answer include the laser strength/scan speed/etc. thanks, Tom



Thomas E. Phillips, Ph.D
Professor of Biological Sciences
Chair, MU Faculty Council
Director, Molecular Cytology Core
2 Tucker Hall
University of Missouri
Columbia, MO 65211-7400
573-882-4712 (office)
573-882-0123 (fax)
[hidden email]



From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of David Knecht
Sent: Friday, December 12, 2008 9:48 AM
To: [hidden email]
Subject: Re: Dictyostelium confocal and heat

Hi Philip- We have been wrestling with this problem for years.  We have heat reflection and heat absorption filters in the transmitted path of all our microscopes.  We always try to minimize light levels for both fluorescence and transmitted as that has a bigger effect than exposure time.  We have used various light sources and all are problematic.  Dicty is a great test of the efficiency of a confocal system, as their rounding up response is so easy to see.  We used that as a diagnostic in evaluating various systems.  In the end, they can all be made to work, but it comes down to how often you take images and how low you can turn down the light and still see your probe.  I think that in general, deconvolution or a laser scanning system in which you can open the pinhole and sacrifice some confocality are the best.  Spinning disk is good, but we just got one so I don't have as much experience yet.  Others have indicated it is an improvement and my initial tests indicate we can image faster.  Our Leica confocal is probably the worst in our stable because you cannot open the pinhole and decrease confocality, however we use it regularly for xyt imaging by just not collecting images as often.  Bottom line- no magic anywhere, only tradeoffs.   Dave

On Dec 12, 2008, at 8:37 AM, Philip Oshel wrote:

Confocalers,

Anyone here working on Dictyostelium? If yes, do you have problems with heat from the illumination source killing the critters?
The cells are really sensitive to heat and light, so I'd like to hear some actual experiences with different systems looking at live Dictys, and how happy they were after you were done looking at them. A regular fluorescence 'scope can easily kill them.

"Illumination source", as I'm wondering about both spinning-disc systems that use Hg or Xe lamps and scanned-laser systems, point scanners or line scanners. We're looking to get a new confocal, and what kind of system (disc, point scanner, line scanner, millions of nanognomes with LED flashlights, whatever) is an issue for the group here.

I'm already dealing with vendors, so no vendor contacts off-list, please. Technical information appropriate for the list is appreciated.

Thanks.

Phil
--
Philip Oshel
Microscopy Facility Supervisor
Biology Department
024C Brooks Hall
Central Michigan University
Mt. Pleasant, MI 48859
(989) 774-3576

Dr. David Knecht    
Department of Molecular and Cell Biology
Co-head Flow Cytometry and Confocal Microscopy Facility
U-3125
91 N. Eagleville Rd.
University of Connecticut
Storrs, CT 06269
860-486-2200
860-486-4331 (fax)

Dr. David Knecht    
Department of Molecular and Cell Biology
Co-head Flow Cytometry and Confocal Microscopy Facility
U-3125
91 N. Eagleville Rd.
University of Connecticut
Storrs, CT 06269
860-486-2200
860-486-4331 (fax)


Oshel, Philip Eugene Oshel, Philip Eugene
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Dictyostelium confocal and heat

Dave,

By "Leica", which one do you mean? A scanned-laser, or your new spinning-disc?

Phil

I don't think this is about raising the temperature of the cells at
all.  We would see a shift in focal plane as the coverslip shifts if
the temp were increasing significantly.  I believe this is more about
photodamage and photooxidation.  If the cells are in HL5 medium which
is very fluorescent, the cells are very sensitive to fluorescent
excitation.  If you change to LoFLo or buffer and allow them to
recycle out their endosomes, they are dramatically less sensitive.
 Developing cells that have been out of medium for 6-8 hours are even
easier to image.  We also found that oxyrase will increase their
resistance to light, indicating photooxidative damage.  I have a time
lapse movie where the cells are imaged on the Leica, where they
rapidly round up and then when you turn the laser down, they very
rapidly begin to extend and move again.  We did some work with
antioxidants, but not enough to find one that solves the problem.
 Dave

On Dec 12, 2008, at 11:39 AM, Phillips, Thomas E. wrote:

David - based on your experience with Dicty, what temperature do you
think your confocals are getting the slides to?  Ideally your answer
include the laser strength/scan speed/etc. thanks, Tom



Thomas E. Phillips, Ph.D
Professor of Biological Sciences
Chair, MU Faculty Council
Director, Molecular Cytology Core
2 Tucker Hall
University of Missouri
Columbia, MO 65211-7400
573-882-4712 (office)
573-882-0123 (fax)
<mailto:[hidden email]>[hidden email]



From: Confocal Microscopy List
[mailto:[hidden email]] On Behalf Of David Knecht
Sent: Friday, December 12, 2008 9:48 AM
To: [hidden email]
Subject: Re: Dictyostelium confocal and heat

Hi Philip- We have been wrestling with this problem for years.  We
have heat reflection and heat absorption filters in the transmitted
path of all our microscopes.  We always try to minimize light levels
for both fluorescence and transmitted as that has a bigger effect
than exposure time.  We have used various light sources and all are
problematic.  Dicty is a great test of the efficiency of a confocal
system, as their rounding up response is so easy to see.  We used
that as a diagnostic in evaluating various systems.  In the end, they
can all be made to work, but it comes down to how often you take
images and how low you can turn down the light and still see your
probe.  I think that in general, deconvolution or a laser scanning
system in which you can open the pinhole and sacrifice some
confocality are the best.  Spinning disk is good, but we just got one
so I don't have as much experience yet.  Others have indicated it is
an improvement and my initial tests indicate we can image faster.
 Our Leica confocal is probably the worst in our stable because you
cannot open the pinhole and decrease confocality, however we use it
regularly for xyt imaging by just not collecting images as often.
 Bottom line- no magic anywhere, only tradeoffs.   Dave

On Dec 12, 2008, at 8:37 AM, Philip Oshel wrote:

Confocalers,

Anyone here working on Dictyostelium? If yes, do you have problems
with heat from the illumination source killing the critters?
The cells are really sensitive to heat and light, so I'd like to hear
some actual experiences with different systems looking at live
Dictys, and how happy they were after you were done looking at them.
A regular fluorescence 'scope can easily kill them.

"Illumination source", as I'm wondering about both spinning-disc
systems that use Hg or Xe lamps and scanned-laser systems, point
scanners or line scanners. We're looking to get a new confocal, and
what kind of system (disc, point scanner, line scanner, millions of
nanognomes with LED flashlights, whatever) is an issue for the group
here.

I'm already dealing with vendors, so no vendor contacts off-list,
please. Technical information appropriate for the list is appreciated.

Thanks.

Phil
--
Philip Oshel
Microscopy Facility Supervisor
Biology Department
024C Brooks Hall
Central Michigan University
Mt. Pleasant, MI 48859
(989) 774-3576

Dr. David Knecht    
Department of Molecular and Cell Biology
Co-head Flow Cytometry and Confocal Microscopy Facility
U-3125
91 N. Eagleville Rd.
University of Connecticut
Storrs, CT 06269
860-486-2200
860-486-4331 (fax)


Dr. David Knecht
Department of Molecular and Cell Biology
Co-head Flow Cytometry and Confocal Microscopy Facility
U-3125
91 N. Eagleville Rd.
University of Connecticut
Storrs, CT 06269
860-486-2200
860-486-4331 (fax)

--
Philip Oshel
Microscopy Facility Supervisor
Biology Department
024C Brooks Hall
Central Michigan University
Mt. Pleasant, MI 48859
(989) 774-3576
Knecht, David Knecht, David
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Dictyostelium confocal and heat

 Leica SP2 laser scanning system.  Our new spinning disk is an Andor system.  Dave

On Dec 12, 2008, at 12:16 PM, Philip Oshel wrote:

Dave,

By "Leica", which one do you mean? A scanned-laser, or your new spinning-disc?

Phil

I don't think this is about raising the temperature of the cells at all.  We would see a shift in focal plane as the coverslip shifts if the temp were increasing significantly.  I believe this is more about photodamage and photooxidation.  If the cells are in HL5 medium which is very fluorescent, the cells are very sensitive to fluorescent excitation.  If you change to LoFLo or buffer and allow them to recycle out their endosomes, they are dramatically less sensitive. Developing cells that have been out of medium for 6-8 hours are even easier to image.  We also found that oxyrase will increase their resistance to light, indicating photooxidative damage.  I have a time lapse movie where the cells are imaged on the Leica, where they rapidly round up and then when you turn the laser down, they very rapidly begin to extend and move again.  We did some work with antioxidants, but not enough to find one that solves the problem. Dave

On Dec 12, 2008, at 11:39 AM, Phillips, Thomas E. wrote:

David - based on your experience with Dicty, what temperature do you think your confocals are getting the slides to?  Ideally your answer include the laser strength/scan speed/etc. thanks, Tom



Thomas E. Phillips, Ph.D
Professor of Biological Sciences
Chair, MU Faculty Council
Director, Molecular Cytology Core
2 Tucker Hall
University of Missouri
Columbia, MO 65211-7400
573-882-4712 (office)
573-882-0123 (fax)
<[hidden email]>[hidden email]



From: Confocal Microscopy List [[hidden email]] On Behalf Of David Knecht
Sent: Friday, December 12, 2008 9:48 AM
To: [hidden email]
Subject: Re: Dictyostelium confocal and heat

Hi Philip- We have been wrestling with this problem for years.  We have heat reflection and heat absorption filters in the transmitted path of all our microscopes.  We always try to minimize light levels for both fluorescence and transmitted as that has a bigger effect than exposure time.  We have used various light sources and all are problematic.  Dicty is a great test of the efficiency of a confocal system, as their rounding up response is so easy to see.  We used that as a diagnostic in evaluating various systems.  In the end, they can all be made to work, but it comes down to how often you take images and how low you can turn down the light and still see your probe.  I think that in general, deconvolution or a laser scanning system in which you can open the pinhole and sacrifice some confocality are the best.  Spinning disk is good, but we just got one so I don't have as much experience yet.  Others have indicated it is an improvement and my initial tests indicate we can image faster. Our Leica confocal is probably the worst in our stable because you cannot open the pinhole and decrease confocality, however we use it regularly for xyt imaging by just not collecting images as often. Bottom line- no magic anywhere, only tradeoffs.   Dave

On Dec 12, 2008, at 8:37 AM, Philip Oshel wrote:

Confocalers,

Anyone here working on Dictyostelium? If yes, do you have problems with heat from the illumination source killing the critters?
The cells are really sensitive to heat and light, so I'd like to hear some actual experiences with different systems looking at live Dictys, and how happy they were after you were done looking at them. A regular fluorescence 'scope can easily kill them.

"Illumination source", as I'm wondering about both spinning-disc systems that use Hg or Xe lamps and scanned-laser systems, point scanners or line scanners. We're looking to get a new confocal, and what kind of system (disc, point scanner, line scanner, millions of nanognomes with LED flashlights, whatever) is an issue for the group here.

I'm already dealing with vendors, so no vendor contacts off-list, please. Technical information appropriate for the list is appreciated.

Thanks.

Phil
--
Philip Oshel
Microscopy Facility Supervisor
Biology Department
024C Brooks Hall
Central Michigan University
Mt. Pleasant, MI 48859
(989) 774-3576

Dr. David Knecht    Department of Molecular and Cell Biology
Co-head Flow Cytometry and Confocal Microscopy Facility
U-3125
91 N. Eagleville Rd.
University of Connecticut
Storrs, CT 06269
860-486-2200
860-486-4331 (fax)


Dr. David Knecht
Department of Molecular and Cell Biology
Co-head Flow Cytometry and Confocal Microscopy Facility
U-3125
91 N. Eagleville Rd.
University of Connecticut
Storrs, CT 06269
860-486-2200
860-486-4331 (fax)

--
Philip Oshel
Microscopy Facility Supervisor
Biology Department
024C Brooks Hall
Central Michigan University
Mt. Pleasant, MI 48859
(989) 774-3576

Dr. David Knecht    
Department of Molecular and Cell Biology
Co-head Flow Cytometry and Confocal Microscopy Facility
U-3125
91 N. Eagleville Rd.
University of Connecticut
Storrs, CT 06269
860-486-2200
860-486-4331 (fax)


George Peeters-2 George Peeters-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Dictyostelium confocal and heat, from a Vendor

In reply to this post by Oshel, Philip Eugene
Dr. John Heuser and Dr. Margret Clark successfully imaged Dicty 6 years ago by using a more sensitive ICCD camera with the Yokogawa CSU-10b. Cameras, spinning disks, and software have improved since then so I do not feel this as a difficult problem to solve. An oxygen membrane overlying the critters was a big help per Dr. Heuser. Perhaps he might comment

I think the old pictures are on our web site.

George



George A. Peeters MD, MS

President,  Solamere Technology Group Inc

1427 Perry Ave

Salt Lake City, UT 84103

www.solameretech.com

801 322-2645 office          801 322-2645 fax

801 232-6911 cell


On Dec 12, 2008, at 6:37 AM, Philip Oshel wrote:

Confocalers,

Anyone here working on Dictyostelium? If yes, do you have problems with heat from the illumination source killing the critters?
The cells are really sensitive to heat and light, so I'd like to hear some actual experiences with different systems looking at live Dictys, and how happy they were after you were done looking at them. A regular fluorescence 'scope can easily kill them.

"Illumination source", as I'm wondering about both spinning-disc systems that use Hg or Xe lamps and scanned-laser systems, point scanners or line scanners. We're looking to get a new confocal, and what kind of system (disc, point scanner, line scanner, millions of nanognomes with LED flashlights, whatever) is an issue for the group here.

I'm already dealing with vendors, so no vendor contacts off-list, please. Technical information appropriate for the list is appreciated.

Thanks.

Phil
--
Philip Oshel
Microscopy Facility Supervisor
Biology Department
024C Brooks Hall
Central Michigan University
Mt. Pleasant, MI 48859
(989) 774-3576

Rietdorf, Jens Rietdorf, Jens
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Dictyostelium confocal and heat

In reply to this post by Oshel, Philip Eugene
Dear Phil,

Ralph Graeph from Univ of Potsdam has compared Yokogawa spinning disk
and single beam confocal microscopy with respect to live imaging
capabilities for Dictyostelium. In these experiments, the laser
intensity was reduced to allow gfp labeled centrosomes to be imaged over
extended periods of time and the contrast achieved with both systems was
compared (Adv Biochem Eng Biotechnol. 2005;95:57-75.) the CSU22 based
system was outcompeting the single beam confocal (a Zeiss LSM 510).
Though it was clearly possible to image live Dicty over extended periods
of time even with both, a single beam scanner and a multi beam scanner
already with the technology available in 2005, recent single beam
scanners have improved in sensitivity, and EMCCD cameras have further
improved the capabilities of spinning disc based systems.

I can ask Ralph for further details of his experiments if required.

Best, jens

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of Philip Oshel
Sent: Friday, December 12, 2008 2:37 PM
To: [hidden email]
Subject: Dictyostelium confocal and heat

Confocalers,

Anyone here working on Dictyostelium? If yes, do you have problems
with heat from the illumination source killing the critters?
The cells are really sensitive to heat and light, so I'd like to hear
some actual experiences with different systems looking at live
Dictys, and how happy they were after you were done looking at them.
A regular fluorescence 'scope can easily kill them.

"Illumination source", as I'm wondering about both spinning-disc
systems that use Hg or Xe lamps and scanned-laser systems, point
scanners or line scanners. We're looking to get a new confocal, and
what kind of system (disc, point scanner, line scanner, millions of
nanognomes with LED flashlights, whatever) is an issue for the group
here.

I'm already dealing with vendors, so no vendor contacts off-list,
please. Technical information appropriate for the list is appreciated.

Thanks.

Phil
--