Hi all,
Would someone please explain the primary differences between Nikon's Lambda series and VC series of Plan Apochromat objectives?

From their website, I understand that the Lambda series objectives have better transmission towards the red end of the spectrum than the VC series objectives. Are there any relative advantages of the VC series?

Thanks!
Kyle

Dr. Kyle M. Douglass
Postdoctoral Fellow
The Laboratory of Experimental Biophysics
EPFL, Lausanne, Switzerland
+41 21 69 30556 (Office)
PEARSON Matthew PEARSON Matthew
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Difference between Nikon Plan Apo Lambda Series and VC Series

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Kyle,

As i understand it the Lambda series lenses only have axial chromatic  
correction down to 435nm whereas the VC (Violet Corrected) have  
chromatic correction down to 405nm.  As you say i think the overall  
transmission across the spectrum is better in the lambda series.

There may well be other differences.

Regards,

Matt










*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

The University of Edinburgh is a charitable body, registered in
Scotland, with registration number SC005336.
Sefic Svara, Diana Sefic Svara, Diana
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Difference between Nikon Plan Apo Lambda Series and VC Series

In reply to this post by Kyle Michael Douglass
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Kyle,

The main difference is in the range of the chromatic aberration corrections. Chromatic aberration is the inability of a lens to focus light of different wavelengths to the same convergence point.
The Lambda series is corrected for blue to far red range 435-850nm while the VC (violet corrected) series corrects from UV to red 405-656nm.

Diana


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Douglass Kyle Michael
Sent: 27 February 2014 15:09
To: [hidden email]
Subject: Difference between Nikon Plan Apo Lambda Series and VC Series

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi all,
Would someone please explain the primary differences between Nikon's Lambda series and VC series of Plan Apochromat objectives?

From their website, I understand that the Lambda series objectives have better transmission towards the red end of the spectrum than the VC series objectives. Are there any relative advantages of the VC series?

Thanks!
Kyle

Dr. Kyle M. Douglass
Postdoctoral Fellow
The Laboratory of Experimental Biophysics EPFL, Lausanne, Switzerland
+41 21 69 30556 (Office)
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Difference between Nikon Plan Apo Lambda Series and VC Series

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I have used the Lambda series 25x 1.1 NA with 375nm excitation and it
actually worked fairly well.  The same lens also works for two-photon all
the way out to 1000 or so as well, as we do two-photon on the same system
(A1) with a Chameleon.

Craig


On Thu, Feb 27, 2014 at 8:37 AM, Sefic Svara, Diana <
[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi Kyle,
>
> The main difference is in the range of the chromatic aberration
> corrections. Chromatic aberration is the inability of a lens to focus light
> of different wavelengths to the same convergence point.
> The Lambda series is corrected for blue to far red range 435-850nm while
> the VC (violet corrected) series corrects from UV to red 405-656nm.
>
> Diana
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]]
> On Behalf Of Douglass Kyle Michael
> Sent: 27 February 2014 15:09
> To: [hidden email]
> Subject: Difference between Nikon Plan Apo Lambda Series and VC Series
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi all,
> Would someone please explain the primary differences between Nikon's
> Lambda series and VC series of Plan Apochromat objectives?
>
> From their website, I understand that the Lambda series objectives have
> better transmission towards the red end of the spectrum than the VC series
> objectives. Are there any relative advantages of the VC series?
>
> Thanks!
> Kyle
>
> Dr. Kyle M. Douglass
> Postdoctoral Fellow
> The Laboratory of Experimental Biophysics EPFL, Lausanne, Switzerland
> +41 21 69 30556 (Office)
>
FLEMING Robert FLEMING Robert
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Difference between Nikon Plan Apo Lambda Series and VC Series

In reply to this post by Kyle Michael Douglass
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Kyle,

The Nikon VC lens will be chromatically corrected at the Violet end (Violet Corrected), we have one on our EC1 confocal and it seems to do what it claims, I don't know how it compares with the lambda lenses,

yours
Bob


Bob Fleming
Bio-Imaging Facility Manager
The Roslin Institute and R(D)SVS
University of Edinburgh
Easter Bush Campus
EH25 9RG
+44 131 651 9409
http://www.roslin.ed.ac.uk/







--
The University of Edinburgh is a charitable body, registered in
Scotland, with registration number SC005336.


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Douglass Kyle Michael
Sent: 27 February 2014 15:09
To: [hidden email]
Subject: Difference between Nikon Plan Apo Lambda Series and VC Series

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi all,
Would someone please explain the primary differences between Nikon's Lambda series and VC series of Plan Apochromat objectives?

From their website, I understand that the Lambda series objectives have better transmission towards the red end of the spectrum than the VC series objectives. Are there any relative advantages of the VC series?

Thanks!
Kyle

Dr. Kyle M. Douglass
Postdoctoral Fellow
The Laboratory of Experimental Biophysics EPFL, Lausanne, Switzerland
+41 21 69 30556 (Office)