Kurt Thorn |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi All - A problem I've run into a few times now is that we have a digital or analog I/O board with relatively high impedance outputs that we want to use to drive a laser with 50 ohm inputs. I've encountered this for both digital and analog control. I'd like to be able to buy a line driver (digital or analog) that I can put in between the I/O board and the laser fix the impedance mismatch. Does anyone know of a source for such a thing? I could build one myself, but I'd rather not if I don't have to. Kurt -- Kurt Thorn Director, Nikon Imaging Center http://nic.ucsf.edu/blog/ |
|
Craig Brideau |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** An emitter-follower circuit should do the trick, depending on how quickly you modulate the laser. Craig On Mon, Feb 24, 2014 at 4:10 PM, Kurt Thorn <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi All - > > A problem I've run into a few times now is that we have a digital or > analog I/O board with relatively high impedance outputs that we want to use > to drive a laser with 50 ohm inputs. I've encountered this for both digital > and analog control. I'd like to be able to buy a line driver (digital or > analog) that I can put in between the I/O board and the laser fix the > impedance mismatch. Does anyone know of a source for such a thing? I could > build one myself, but I'd rather not if I don't have to. > > Kurt > > -- > Kurt Thorn > Director, Nikon Imaging Center > http://nic.ucsf.edu/blog/ > |
|
Craig Brideau |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Here's a calculator: http://www.ece.umd.edu/~neil/306/ckts/ef.htm And a little tutorial: http://www.eecs.tufts.edu/~dsculley/tutorial/transistors/transistors2.html Note they are also known as common collector circuits. Basically it isolates one section of a circuit from another. Pretty handy! Craig On Mon, Feb 24, 2014 at 4:34 PM, Craig Brideau <[hidden email]>wrote: > An emitter-follower circuit should do the trick, depending on how quickly > you modulate the laser. > > Craig > > > > On Mon, Feb 24, 2014 at 4:10 PM, Kurt Thorn <[hidden email]> wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hi All - >> >> A problem I've run into a few times now is that we have a digital or >> analog I/O board with relatively high impedance outputs that we want to use >> to drive a laser with 50 ohm inputs. I've encountered this for both digital >> and analog control. I'd like to be able to buy a line driver (digital or >> analog) that I can put in between the I/O board and the laser fix the >> impedance mismatch. Does anyone know of a source for such a thing? I could >> build one myself, but I'd rather not if I don't have to. >> >> Kurt >> >> -- >> Kurt Thorn >> Director, Nikon Imaging Center >> http://nic.ucsf.edu/blog/ >> > > |
|
Mark Cannell-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Yes, but remember the one diode drop difference (Vbe) between input and output (not a problem for digital). For digital such things are known as line drivers and you can get 4 (or more) in one package. For analog, something like an AD811 should do. I don’t know of a ready built solution for this interface problem, if you can’t make the circuit, there are other computer interface boards that will drive 50 ohm loads that may provide a simpler solution for you… HTH Mark > > > > On Mon, Feb 24, 2014 at 4:34 PM, Craig Brideau <[hidden email]>wrote: > >> An emitter-follower circuit should do the trick, depending on how quickly >> you modulate the laser. >> >> Craig >> >> >> >> On Mon, Feb 24, 2014 at 4:10 PM, Kurt Thorn <[hidden email]> wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Hi All - >>> >>> A problem I've run into a few times now is that we have a digital or >>> analog I/O board with relatively high impedance outputs that we want to use >>> to drive a laser with 50 ohm inputs. I've encountered this for both digital >>> and analog control. I'd like to be able to buy a line driver (digital or >>> analog) that I can put in between the I/O board and the laser fix the >>> impedance mismatch. Does anyone know of a source for such a thing? I could >>> build one myself, but I'd rather not if I don't have to. >>> >>> Kurt >>> >>> -- >>> Kurt Thorn >>> Director, Nikon Imaging Center >>> http://nic.ucsf.edu/blog/ >>> >> >> Mark B. Cannell Ph.D. FRSNZ Professor of Cardiac Cell Biology School of Physiology & Pharmacology Medical Sciences Building University of Bristol Bristol BS8 1TD UK [hidden email] |
|
Andrew York |
|
|
In reply to this post by Kurt Thorn
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I've run into this problem myself, and was surprised I couldn't find a simple premade buffer amplifier for sale. I ended up building my own with radio shack parts, but it's annoying for sure. Some NI boards can drive low impedance loads, I think the 6733 works. Some are spec'd to do this, but fail. I'm interested in a good approach to this problem too, if anyone can help. On Feb 24, 2014 6:11 PM, "Kurt Thorn" <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi All - > > A problem I've run into a few times now is that we have a digital or > analog I/O board with relatively high impedance outputs that we want to use > to drive a laser with 50 ohm inputs. I've encountered this for both digital > and analog control. I'd like to be able to buy a line driver (digital or > analog) that I can put in between the I/O board and the laser fix the > impedance mismatch. Does anyone know of a source for such a thing? I could > build one myself, but I'd rather not if I don't have to. > > Kurt > > -- > Kurt Thorn > Director, Nikon Imaging Center > http://nic.ucsf.edu/blog/ > |
|
Zdenek Svindrych |
|
|
In reply to this post by Kurt Thorn
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Kurt, have you checked the specs of your I/O board? High impedance analog outputs are rare on good quality lab equipment and high impedance digital outputs are a mere nonsense. It is possible that your board uses 3.3V interface, this is common on modern cheapo hardware. You cancheck this with an oscilloscope (or voltmeter) andsome dummy load. Just check whether there are some hardware jumpers or software settings that could swithch the outputs to low impedance. The same may apply for your laser. High impedance (about 1 kOhm) is more common, the 50 Ohm is an option that might be "deactivated". Again look into some detailed manual. It is simple to build a level converter or a buffer yourself. I couldn't find any simple ready solution for you on the net... Most of the results (both devices and schematics) I could find are impedance converters for very high frequencies, that's an overkill. Good luck! zdenek svindrych 1st faculty of medicine, cuni.cz ---------- Původní zpráva ---------- Od: Kurt Thorn <[hidden email]> Komu: [hidden email] Datum: 25. 2. 2014 0:11:12 Předmět: Digitial and Analog line drivers for laser control "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi All - A problem I've run into a few times now is that we have a digital or analog I/O board with relatively high impedance outputs that we want to use to drive a laser with 50 ohm inputs. I've encountered this for both digital and analog control. I'd like to be able to buy a line driver (digital or analog) that I can put in between the I/O board and the laser fix the impedance mismatch. Does anyone know of a source for such a thing? I could build one myself, but I'd rather not if I don't have to. Kurt -- Kurt Thorn Director, Nikon Imaging Center http://nic.ucsf.edu/blog/" |