Double TSA?

classic Classic list List threaded Threaded
♦
♦
6 messages Options
Richard Green-2 Richard Green-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Double TSA?

So I have recently been working on a double staining IHC using TSA to  
amplify the signal in one antibody. It's been working great except  
that as s result of all the washing steps from the TSA protocol the  
other non TSA antibody stain is too weak now. I have tried to decrease  
the washing steps which hadn't helped that much. Any suggestions on  
how I can apply TSA/ amplification to two antibodies? I have
Looked around online and haven't found anything concrete. Any  
suggestions folks have is muchly appreciated. This is using parafin  
embedded adipose tissue. Thanks!
Richard Green
Goodhouse, Joseph G. Goodhouse, Joseph G.
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Double TSA?

        It could be an affinity problem with the first anti-body.  One
trick to solve this is to re-fix the sample after you  have done your
first primary anti-body incubation and washes to cross link it to its
antigen.

Joe Goodhouse
Confocal Core Lab Manager
Dept. of Molecular Biology
Princeton University
Washington Road
Princeton, NJ. 08544-1014
609-258-5432
Visit us at http://www.molbio1.princeton.edu/facility/confocal/


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of Richard Green
Sent: Tuesday, December 01, 2009 3:28 PM
To: [hidden email]
Subject: Double TSA?

So I have recently been working on a double staining IHC using TSA to  
amplify the signal in one antibody. It's been working great except  
that as s result of all the washing steps from the TSA protocol the  
other non TSA antibody stain is too weak now. I have tried to decrease  
the washing steps which hadn't helped that much. Any suggestions on  
how I can apply TSA/ amplification to two antibodies? I have
Looked around online and haven't found anything concrete. Any  
suggestions folks have is muchly appreciated. This is using parafin  
embedded adipose tissue. Thanks!
Richard Green
Cameron Nowell Cameron Nowell
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Double TSA?

Hi Richard,

There are two variations of TSA, HRP and AP. I would think if you use
HRP for one antibody and AP for the other you should be able to double
label.

Cheers


Cam



Cameron J. Nowell
Microscopy Manager
Centre for Advanced Microscopy
Ludwig Institute for Cancer Research
PO Box 2008
Royal Melbourne Hospital
Victoria, 3050
AUSTRALIA
Office: +61 3 9341 3155
Mobile: +61422882700
Fax: +61 3 9341 3104
Facility Website



-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of Goodhouse, Joseph G.
Sent: Wednesday, 2 December 2009 7:36 AM
To: [hidden email]
Subject: Re: Double TSA?

        It could be an affinity problem with the first anti-body.  One
trick to solve this is to re-fix the sample after you  have done your
first primary anti-body incubation and washes to cross link it to its
antigen.

Joe Goodhouse
Confocal Core Lab Manager
Dept. of Molecular Biology
Princeton University
Washington Road
Princeton, NJ. 08544-1014
609-258-5432
Visit us at http://www.molbio1.princeton.edu/facility/confocal/


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of Richard Green
Sent: Tuesday, December 01, 2009 3:28 PM
To: [hidden email]
Subject: Double TSA?

So I have recently been working on a double staining IHC using TSA to  
amplify the signal in one antibody. It's been working great except  
that as s result of all the washing steps from the TSA protocol the  
other non TSA antibody stain is too weak now. I have tried to decrease  
the washing steps which hadn't helped that much. Any suggestions on  
how I can apply TSA/ amplification to two antibodies? I have
Looked around online and haven't found anything concrete. Any  
suggestions folks have is muchly appreciated. This is using parafin  
embedded adipose tissue. Thanks!
Richard Green

No virus found in this incoming message.
Checked by AVG - www.avg.com
Version: 9.0.709 / Virus Database: 270.14.88/2537 - Release Date:
12/02/09 06:32:00


This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research Ltd does not waive any rights if you have received this communication in error.
The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research Ltd.

Asson-Batres, Mary Ann Asson-Batres, Mary Ann
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Double TSA?

In reply to this post by Richard Green-2
We published a method for double labeling with TSA and continue to use it with success for cultured cells and different tissues fixed with paraformaldehyde and embedded in paraffin or OCT-frozen sections.  You can find it in this reference:

J Comp Neurol. 2006 May 10;496(2):149-71.  Localization of retinaldehyde dehydrogenases and retinoid binding proteins to sustentacular cells, glia, Bowman's gland cells, and stroma: potential sites of retinoic acid synthesis in the postnatal rat olfactory organ.

at http://www.ncbi.nlm.nih.gov/pubmed/16538685?itool=EntrezSystem2.PEntrez.Pubmed.Pubmed_ResultsPanel.Pubmed_RVDocSum&ordinalpos=2


Mary Ann

Mary Ann Asson-Batres, PhD
Professor
Department of Biological Sciences
Tennessee State University
PO Box 1116 (postal mail)
Room 313 Harned Hall (courier)
3500 John A Merritt Blvd
Nashville, TN  37209

Office  615-963-5779
FAX    615-963-2142

[hidden email]
[hidden email]


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Richard Green
Sent: Tuesday, December 01, 2009 2:28 PM
To: [hidden email]
Subject: Double TSA?

So I have recently been working on a double staining IHC using TSA to amplify the signal in one antibody. It's been working great except that as s result of all the washing steps from the TSA protocol the other non TSA antibody stain is too weak now. I have tried to decrease the washing steps which hadn't helped that much. Any suggestions on how I can apply TSA/ amplification to two antibodies? I have Looked around online and haven't found anything concrete. Any suggestions folks have is muchly appreciated. This is using parafin embedded adipose tissue. Thanks!
Richard Green
Tamara Howard Tamara Howard
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Double TSA?

In reply to this post by Richard Green-2
I haven't tried this protocol but have a reference from
Toth & Mezey in 2007; they were trying to do
double-labeling with primaries from
the same species & ended up detecting with TSA: J.
Histochem. & Cytochem. (55) 6: 545-554

Tamara



On Tue, 1 Dec 2009 12:28:14 -0800
  Richard Green <[hidden email]> wrote:

> So I have recently been working on a double staining IHC
>using TSA to  amplify the signal in one antibody. It's
>been working great except  that as s result of all the
>washing steps from the TSA protocol the  other non TSA
>antibody stain is too weak now. I have tried to decrease
> the washing steps which hadn't helped that much. Any
>suggestions on  how I can apply TSA/ amplification to two
>antibodies? I have
> Looked around online and haven't found anything
>concrete. Any  suggestions folks have is muchly
>appreciated. This is using parafin  embedded adipose
>tissue. Thanks!
> Richard Green

***************************
Tamara Howard
Cell Biology & Physiology
UNM-HSC
Albuquerque, NM
***************************
Francisco J H Blazquez Francisco J H Blazquez
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Double TSA?

Hello..I had the same problem some years ago, so we did a double amplification using first a biotinilated-FITC marker amplified by an anti-FITC-ALEXA 433 antibody, then we did the TSA with TRITC as fluorochrome.

If you wish the method is described here and you may see the photos of the double staining

<a href="http://dx.doi.org/10.1006/excr.2001.5342" target="doilink" onclick="var doiWin; doiWin=window.open('http://dx.doi.org/10.1006/excr.2001.5342','doilink','scrollbars=yes,resizable=yes,directories=yes,toolbar=yes,menubar=yes,status=yes'); doiWin.focus()">doi:10.1006/excr.2001.5342 
Experimental Cell Research
Volume 270, Issue 2, 1 November 2001, Pages 235-247

Good luck

Prof. Dr. Francisco Javier Hernandez Blazquez
University of São Paulo
School of Veterinary Medicine
Department of Surgery
Av. Prof. Dr. Orlando Marques de Paiva, 87
05508-270 - São Paulo (SP) - Brasil
http://www.fmvz.usp.br/index.php/site/docentes/lista_de_docentes/francisco_javier_hernandez_blazquez
Tel..55 (11) 3091 1374  Fax  55 (11) 3091 7805
email: [hidden email]


Tamara A Howard escreveu:
I haven't tried this protocol but have a reference from Toth & Mezey in 2007; they were trying to do double-labeling with primaries from
the same species & ended up detecting with TSA: J. Histochem. & Cytochem. (55) 6: 545-554

Tamara



On Tue, 1 Dec 2009 12:28:14 -0800
 Richard Green [hidden email] wrote:
So I have recently been working on a double staining IHC using TSA to  amplify the signal in one antibody. It's been working great except  that as s result of all the washing steps from the TSA protocol the  other non TSA antibody stain is too weak now. I have tried to decrease the washing steps which hadn't helped that much. Any suggestions on  how I can apply TSA/ amplification to two antibodies? I have
Looked around online and haven't found anything concrete. Any  suggestions folks have is muchly appreciated. This is using parafin  embedded adipose tissue. Thanks!
Richard Green

***************************
Tamara Howard
Cell Biology & Physiology
UNM-HSC
Albuquerque, NM
***************************






--