Drift correction for live cell imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi all,

I tracking a single spot inside the nucleus for live cell imaging ( 50 sec duration) by spinning disk microscopy. Does anyone has suggestions for correcting the drift (lateral) in this case?

Thanks,
Haitham


Haitham A. Shaban
Chromatin and Gene Expression Group
Laboratoire de Biologie Moléculaire Eucaryote - CNRS
University of Toulouse
Jonathan Chubb Jonathan Chubb
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Drift correction for live cell imaging

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

If you have a second spot, then you can use that as a reference, such as another locus- or the same locus in a diploid (see Wallace Marshall/John Sedat- Current Biology 1997, our paper in 2002).  Some people use the SPB.  Susan Gasser had a paper in 2001 (Heun et al, Science) where they didn't have a reference spot, but their spots were moving so much (yeast), it really didn't matter.


There is probably some fancy Fourier way to decompose the drift and diffusion components from the time series data, but I am not the person to ask.




From: Confocal Microscopy List <[hidden email]> on behalf of Haitham Shaban <[hidden email]>
Sent: 06 October 2016 20:33
To: [hidden email]
Subject: Drift correction for live cell imaging
 
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
lists.umn.edu
[hidden email]: listserv archives. confocalmicroscopy


Post images on http://www.imgur.com and include the link in your posting.
www.imgur.com
Imgur is the best place to share and enjoy the most awesome images on the Internet. Every day, millions of people use Imgur to be entertained and inspired by funny ...


*****

Hi all,

I tracking a single spot inside the nucleus for live cell imaging ( 50 sec duration) by spinning disk microscopy. Does anyone has suggestions for correcting the drift (lateral) in this case?

Thanks,
Haitham


Haitham A. Shaban
Chromatin and Gene Expression Group
Laboratoire de Biologie Moléculaire Eucaryote - CNRS
University of Toulouse
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Drift correction for live cell imaging

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
XY drift is pretty straightforward. Do a rolling convolution on the images and note the offsets at which you get maximum correlation. That will be your offset between frames.

Craig

On Thu, Oct 6, 2016 at 1:52 PM, Chubb, Jonathan <[hidden email]> wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

If you have a second spot, then you can use that as a reference, such as another locus- or the same locus in a diploid (see Wallace Marshall/John Sedat- Current Biology 1997, our paper in 2002).  Some people use the SPB.  Susan Gasser had a paper in 2001 (Heun et al, Science) where they didn't have a reference spot, but their spots were moving so much (yeast), it really didn't matter.


There is probably some fancy Fourier way to decompose the drift and diffusion components from the time series data, but I am not the person to ask.




From: Confocal Microscopy List <[hidden email]> on behalf of Haitham Shaban <[hidden email]>
Sent: 06 October 2016 20:33
To: [hidden email]
Subject: Drift correction for live cell imaging
 
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy


Post images on http://www.imgur.com and include the link in your posting.
Imgur is the best place to share and enjoy the most awesome images on the Internet. Every day, millions of people use Imgur to be entertained and inspired by funny ...


*****

Hi all,

I tracking a single spot inside the nucleus for live cell imaging ( 50 sec duration) by spinning disk microscopy. Does anyone has suggestions for correcting the drift (lateral) in this case?

Thanks,
Haitham


Haitham A. Shaban
Chromatin and Gene Expression Group
Laboratoire de Biologie Moléculaire Eucaryote - CNRS
University of Toulouse

Joshua Zachary Rappoport Joshua Zachary Rappoport
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Drift correction for live cell imaging

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

We wrote software to do this automatically (see https://www.ncbi.nlm.nih.gov/pubmed/22044432)

Email me directly if you want us to share it with you

 

Josh

 

Joshua Z. Rappoport PhD

 

Director of the Center for Advanced Microscopy (CAM) and Nikon Imaging Center (NIC)

Northwestern University Feinberg School of Medicine
303 E. Chicago Avenue
Chicago, IL 60611

(312) 503-4140

http://cam.facilities.northwestern.edu/

http://nic.feinberg.northwestern.edu/

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau
Sent: Thursday, October 06, 2016 4:29 PM
To: [hidden email]
Subject: Re: Drift correction for live cell imaging

 

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

XY drift is pretty straightforward. Do a rolling convolution on the images and note the offsets at which you get maximum correlation. That will be your offset between frames.

 

Craig

 

On Thu, Oct 6, 2016 at 1:52 PM, Chubb, Jonathan <[hidden email]> wrote:

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

If you have a second spot, then you can use that as a reference, such as another locus- or the same locus in a diploid (see Wallace Marshall/John Sedat- Current Biology 1997, our paper in 2002).  Some people use the SPB.  Susan Gasser had a paper in 2001 (Heun et al, Science) where they didn't have a reference spot, but their spots were moving so much (yeast), it really didn't matter.

 

There is probably some fancy Fourier way to decompose the drift and diffusion components from the time series data, but I am not the person to ask.

 


From: Confocal Microscopy List <[hidden email]> on behalf of Haitham Shaban <[hidden email]>
Sent: 06 October 2016 20:33
To: [hidden email]
Subject: Drift correction for live cell imaging

 

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy



Post images on http://www.imgur.com and include the link in your posting.

Imgur is the best place to share and enjoy the most awesome images on the Internet. Every day, millions of people use Imgur to be entertained and inspired by funny ...



*****

Hi all,

I tracking a single spot inside the nucleus for live cell imaging ( 50 sec duration) by spinning disk microscopy. Does anyone has suggestions for correcting the drift (lateral) in this case?

Thanks,
Haitham


Haitham A. Shaban
Chromatin and Gene Expression Group
Laboratoire de Biologie Moléculaire Eucaryote - CNRS
University of Toulouse

 

G. Esteban Fernandez G. Esteban Fernandez
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Drift correction for live cell imaging

In reply to this post by Haitham Shaban
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I use the following functions in FIJI ImageJ (free at http://fiji.sc):

For XY drift:  Plugins > Registration > StackReg
For XYZ drift:  Plugins > Registration > Correct 3D drift

-Esteban

On Thu, Oct 6, 2016 at 12:33 PM, Haitham Shaban
<[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi all,
>
> I tracking a single spot inside the nucleus for live cell imaging ( 50 sec duration) by spinning disk microscopy. Does anyone has suggestions for correcting the drift (lateral) in this case?
>
> Thanks,
> Haitham
>
>
> Haitham A. Shaban
> Chromatin and Gene Expression Group
> Laboratoire de Biologie Moléculaire Eucaryote - CNRS
> University of Toulouse
Haitham Shaban Haitham Shaban
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Drift correction for live cell imaging

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Thanks Esteban for your answer.
I think using StackReg Plugin for live cell imaging correction is incorrect, because it uses each slice as a template with respect to which the next slice, and then the correction goes in one direction. Since every thing inside the recorded video is moving so the correction by using StackReg will add wrong movement on the spot of interest for tracking. 

Best,
Haitham

On Fri, Oct 7, 2016 at 12:20 AM, G. Esteban Fernandez <[hidden email]> wrote:
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I use the following functions in FIJI ImageJ (free at http://fiji.sc):

For XY drift:  Plugins > Registration > StackReg
For XYZ drift:  Plugins > Registration > Correct 3D drift

-Esteban

On Thu, Oct 6, 2016 at 12:33 PM, Haitham Shaban
<[hidden email]> wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi all,
>
> I tracking a single spot inside the nucleus for live cell imaging ( 50 sec duration) by spinning disk microscopy. Does anyone has suggestions for correcting the drift (lateral) in this case?
>
> Thanks,
> Haitham
>
>
> Haitham A. Shaban
> Chromatin and Gene Expression Group
> Laboratoire de Biologie Moléculaire Eucaryote - CNRS
> University of Toulouse

TSwayne TSwayne
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Drift correction for live cell imaging

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Hi Haitham,

As Jonathan mentioned, you need some other recognizable structure to use as a fiducial.  Do you have other channels in your image? 

If you have only one channel, maybe there is enough background labeling to segment the nucleus.  You could then use that mask as a basis for automatic alignment — that way, the plugin will be looking at the whole nucleus and will not try to register the moving spot. 

I believe you could then save the shift values and apply them to your original movie, so that you can track the spot relative to the nucleus.

If you could post an example image or short sequence it might inspire other solutions.

Hope this helps.

------------------------------------
Theresa Swayne, Ph.D.
Manager

Herbert Irving Comprehensive Cancer Center
Columbia University Medical Center
1130 St. Nicholas Ave., Room 222A
New York, NY 10032
Phone: 212-851-4613


On Oct 7, 2016, at 12:10 AM, Haitham Shaban <[hidden email]> wrote:

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Thanks Esteban for your answer.
I think using StackReg Plugin for live cell imaging correction is incorrect, because it uses each slice as a template with respect to which the next slice, and then the correction goes in one direction. Since every thing inside the recorded video is moving so the correction by using StackReg will add wrong movement on the spot of interest for tracking. 

Best,
Haitham

On Fri, Oct 7, 2016 at 12:20 AM, G. Esteban Fernandez <[hidden email]> wrote:
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I use the following functions in FIJI ImageJ (free at http://fiji.sc):

For XY drift:  Plugins > Registration > StackReg
For XYZ drift:  Plugins > Registration > Correct 3D drift

-Esteban

On Thu, Oct 6, 2016 at 12:33 PM, Haitham Shaban
<[hidden email]> wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi all,
>
> I tracking a single spot inside the nucleus for live cell imaging ( 50 sec duration) by spinning disk microscopy. Does anyone has suggestions for correcting the drift (lateral) in this case?
>
> Thanks,
> Haitham
>
>
> Haitham A. Shaban
> Chromatin and Gene Expression Group
> Laboratoire de Biologie Moléculaire Eucaryote - CNRS
> University of Toulouse



Haitham Shaban Haitham Shaban
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Drift correction for live cell imaging

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Hi Theresa,

Thanks for your helpful suggestion.I think the main problem is that I do not have anything is fixed in my image stack. I have background around the nucleus that could be used as a fiducial, but since it is a part of the cell so it also has movement. Also it suppose that both of them have different motion behaviour so if we went to relative the motions for the spot based on the mask came out from the background, it will add artifact on the motion of the spot. Please correct me if I am wrong.

I found it is forbidden to attach images and videos within this group; please find download a small video for this link spot of my interest inside the nucleus, that is could be good example for real problem :)

https://www.sendbigfiles.com/76e07de679d73aaca015d6d13602e3a9/

Best regards,
Haitham

On Fri, Oct 7, 2016 at 4:43 PM, Haitham Shaban <[hidden email]> wrote:
Hi Theresa,

Thanks for your helpful suggestion.I think the main problem is that I do not have anything is fixed in my image stack. I have background around the nucleus that could be used as a fiducial, but since it is a part of the cell so it also has movement. Also it suppose that both of them have different motion behaviour so if we went to relative the motions for the spot based on the mask came out from the background, it will add artifact on the motion of the spot. Please correct me if I am wrong.

Please find enclosed a small video for the spot of my interest inside the nucleus, that is could be good example for real problem :)

Best regards,
Haitham

On Fri, Oct 7, 2016 at 3:54 PM, Swayne, Theresa C. <[hidden email]> wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Hi Haitham,

As Jonathan mentioned, you need some other recognizable structure to use as a fiducial.  Do you have other channels in your image? 

If you have only one channel, maybe there is enough background labeling to segment the nucleus.  You could then use that mask as a basis for automatic alignment — that way, the plugin will be looking at the whole nucleus and will not try to register the moving spot. 

I believe you could then save the shift values and apply them to your original movie, so that you can track the spot relative to the nucleus.

If you could post an example image or short sequence it might inspire other solutions.

Hope this helps.

------------------------------------
Theresa Swayne, Ph.D.
Manager

Herbert Irving Comprehensive Cancer Center
Columbia University Medical Center
1130 St. Nicholas Ave., Room 222A
New York, NY 10032
Phone: 212-851-4613


On Oct 7, 2016, at 12:10 AM, Haitham Shaban <[hidden email]> wrote:

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Thanks Esteban for your answer.
I think using StackReg Plugin for live cell imaging correction is incorrect, because it uses each slice as a template with respect to which the next slice, and then the correction goes in one direction. Since every thing inside the recorded video is moving so the correction by using StackReg will add wrong movement on the spot of interest for tracking. 

Best,
Haitham

On Fri, Oct 7, 2016 at 12:20 AM, G. Esteban Fernandez <[hidden email]> wrote:
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I use the following functions in FIJI ImageJ (free at http://fiji.sc):

For XY drift:  Plugins > Registration > StackReg
For XYZ drift:  Plugins > Registration > Correct 3D drift

-Esteban

On Thu, Oct 6, 2016 at 12:33 PM, Haitham Shaban
<[hidden email]> wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi all,
>
> I tracking a single spot inside the nucleus for live cell imaging ( 50 sec duration) by spinning disk microscopy. Does anyone has suggestions for correcting the drift (lateral) in this case?
>
> Thanks,
> Haitham
>
>
> Haitham A. Shaban
> Chromatin and Gene Expression Group
> Laboratoire de Biologie Moléculaire Eucaryote - CNRS
> University of Toulouse





Julio Vazquez-2 Julio Vazquez-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Drift correction for live cell imaging

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Motion is only meaningful relative to a reference system. If you are tracking a single spot inside the nucleus, what is the reference system you are measuring the spot relative to? the slide may be moving relative to the camera; the cell may be moving relative to the slide; the nucleus may be moving relative to the cell... So I would say you first need to define precisely what motion you want to measure, and then choose and label a suitable reference object so that you can make your measurements. We faced a similar situation a few years ago when measuring chromosome motion, and our take was similar to that already mentioned of Wallace Marshall et al, namely to label two spots, and track their relative distance over time (Current Biology 11(16): 1227-1239. 2001). This gave us a measurement of the relative motion of two chromatin loci inside the nucleus, and was independent of the motion of everything else, including stage drift. If we assume that the two spots behave in a similar manner, then we can deduce the motion of each spot (or at  least some aspects of it). We were lucky  that the cells we were studying didn't have much motion themselves, so we could get a pretty good idea of the motion of each spot and verify that the two spots behaved in a very similar manner, but this is not always possible. Therefore, you need to take some time to think about the biological question you are interested in, which will lead you to a possible solution. For example, you could find the center of the nucleus, and use that as the reference (i.e. measure the distance between your spot and the center of the nucleus); or you could measure the shortest distance between your spot and the nuclear boundary; and so on.

Julio Vazquez
Fred Hutchinson Cancer Research Center
--

On Oct 7, 2016, at 7:52 AM, Haitham Shaban wrote:

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Hi Theresa,

Thanks for your helpful suggestion.I think the main problem is that I do not have anything is fixed in my image stack. I have background around the nucleus that could be used as a fiducial, but since it is a part of the cell so it also has movement. Also it suppose that both of them have different motion behaviour so if we went to relative the motions for the spot based on the mask came out from the background, it will add artifact on the motion of the spot. Please correct me if I am wrong.

I found it is forbidden to attach images and videos within this group; please find download a small video for this link spot of my interest inside the nucleus, that is could be good example for real problem :)

https://www.sendbigfiles.com/76e07de679d73aaca015d6d13602e3a9/

Best regards,
Haitham

On Fri, Oct 7, 2016 at 4:43 PM, Haitham Shaban <[hidden email]> wrote:
Hi Theresa,

Thanks for your helpful suggestion.I think the main problem is that I do not have anything is fixed in my image stack. I have background around the nucleus that could be used as a fiducial, but since it is a part of the cell so it also has movement. Also it suppose that both of them have different motion behaviour so if we went to relative the motions for the spot based on the mask came out from the background, it will add artifact on the motion of the spot. Please correct me if I am wrong.

Please find enclosed a small video for the spot of my interest inside the nucleus, that is could be good example for real problem :)

Best regards,
Haitham

On Fri, Oct 7, 2016 at 3:54 PM, Swayne, Theresa C. <[hidden email]> wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Hi Haitham,

As Jonathan mentioned, you need some other recognizable structure to use as a fiducial.  Do you have other channels in your image? 

If you have only one channel, maybe there is enough background labeling to segment the nucleus.  You could then use that mask as a basis for automatic alignment — that way, the plugin will be looking at the whole nucleus and will not try to register the moving spot. 

I believe you could then save the shift values and apply them to your original movie, so that you can track the spot relative to the nucleus.

If you could post an example image or short sequence it might inspire other solutions.

Hope this helps.

------------------------------------
Theresa Swayne, Ph.D.
Manager

Herbert Irving Comprehensive Cancer Center
Columbia University Medical Center
1130 St. Nicholas Ave., Room 222A
New York, NY 10032
Phone: 212-851-4613


On Oct 7, 2016, at 12:10 AM, Haitham Shaban <[hidden email]> wrote:

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Thanks Esteban for your answer.
I think using StackReg Plugin for live cell imaging correction is incorrect, because it uses each slice as a template with respect to which the next slice, and then the correction goes in one direction. Since every thing inside the recorded video is moving so the correction by using StackReg will add wrong movement on the spot of interest for tracking. 

Best,
Haitham

On Fri, Oct 7, 2016 at 12:20 AM, G. Esteban Fernandez <[hidden email]> wrote:
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I use the following functions in FIJI ImageJ (free at http://fiji.sc):

For XY drift:  Plugins > Registration > StackReg
For XYZ drift:  Plugins > Registration > Correct 3D drift

-Esteban

On Thu, Oct 6, 2016 at 12:33 PM, Haitham Shaban
<[hidden email]> wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi all,
>
> I tracking a single spot inside the nucleus for live cell imaging ( 50 sec duration) by spinning disk microscopy. Does anyone has suggestions for correcting the drift (lateral) in this case?
>
> Thanks,
> Haitham
>
>
> Haitham A. Shaban
> Chromatin and Gene Expression Group
> Laboratoire de Biologie Moléculaire Eucaryote - CNRS
> University of Toulouse






TSwayne TSwayne
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Drift correction for live cell imaging

In reply to this post by Haitham Shaban
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Hi Haitham,

I agree wholeheartedly with Julio Vazquez’s comments (included below). Your first task is to decide what frame of reference you will use to measure the motion of the spot. Nucleus, cell membrane, cell centroid, etc.  It does not have to be a fixed point, it just needs to be a feature that you can identify in every frame. 

The purpose of drift correction/registration is to align that feature in every timepoint so that it *becomes* fixed. Then you can measure the motion of other things in the movie relative to that feature. I hope that makes sense.

Looking at the sample image ex.tif, I see a bright moving nuclear spot, but I don’t see any drift of the whole nucleus or cell. So I’m not sure what you want to correct for in that movie.

Best,
Theresa


On Oct 7, 2016, at 10:52 AM, Haitham Shaban <[hidden email]> wrote:

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Hi Theresa,

Thanks for your helpful suggestion.I think the main problem is that I do not have anything is fixed in my image stack. I have background around the nucleus that could be used as a fiducial, but since it is a part of the cell so it also has movement. Also it suppose that both of them have different motion behaviour so if we went to relative the motions for the spot based on the mask came out from the background, it will add artifact on the motion of the spot. Please correct me if I am wrong.

I found it is forbidden to attach images and videos within this group; please find download a small video for this link spot of my interest inside the nucleus, that is could be good example for real problem :)

https://www.sendbigfiles.com/76e07de679d73aaca015d6d13602e3a9/

Best regards,
Haitham



On Oct 7, 2016, at 11:58 AM, Julio Vazquez <[hidden email]> wrote:

Motion is only meaningful relative to a reference system. If you are tracking a single spot inside the nucleus, what is the reference system you are measuring the spot relative to? the slide may be moving relative to the camera; the cell may be moving relative to the slide; the nucleus may be moving relative to the cell... So I would say you first need to define precisely what motion you want to measure, and then choose and label a suitable reference object so that you can make your measurements. We faced a similar situation a few years ago when measuring chromosome motion, and our take was similar to that already mentioned of Wallace Marshall et al, namely to label two spots, and track their relative distance over time (Current Biology 11(16): 1227-1239. 2001). This gave us a measurement of the relative motion of two chromatin loci inside the nucleus, and was independent of the motion of everything else, including stage drift. If we assume that the two spots behave in a similar manner, then we can deduce the motion of each spot (or at  least some aspects of it). We were lucky  that the cells we were studying didn't have much motion themselves, so we could get a pretty good idea of the motion of each spot and verify that the two spots behaved in a very similar manner, but this is not always possible. Therefore, you need to take some time to think about the biological question you are interested in, which will lead you to a possible solution. For example, you could find the center of the nucleus, and use that as the reference (i.e. measure the distance between your spot and the center of the nucleus); or you could measure the shortest distance between your spot and the nuclear boundary; and so on.

------------------------------------
Theresa Swayne, Ph.D.
Manager

Herbert Irving Comprehensive Cancer Center
Columbia University Medical Center
1130 St. Nicholas Ave., Room 222A
New York, NY 10032
Phone: 212-851-4613

mcammer mcammer
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Drift correction for live cell imaging

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

An extreme point about alignment over time.

 

https://www.flickr.com/photos/mcammer/3623501576/ This movie is an intermediate step in aligning a sperm that was swimming in a circle; the subsequent step (don’t know where the movie is) aligned the sperm head so wasn’t moving at all, only tail was, and this was used to measure the tail movement.

 

The image on the top https://www.flickr.com/photos/mcammer/3623501756/ is a projection of the original movie and the image on the bottom is a projection of the aligned image.

 

=========================================================================

Michael Cammer, Microscopy Core & Skirball Institute, NYU Langone Medical Center

         Cell:  914-309-3270     Office: Skirball 2nd Floor main office, back right

            http://ocs.med.nyu.edu/microscopy & http://microscopynotes.com/

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Swayne, Theresa C.
Sent: Monday, October 10, 2016 2:45 PM
To: [hidden email]
Subject: Re: Drift correction for live cell imaging

 

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

 

Hi Haitham,

 

I agree wholeheartedly with Julio Vazquez’s comments (included below). Your first task is to decide what frame of reference you will use to measure the motion of the spot. Nucleus, cell membrane, cell centroid, etc.  It does not have to be a fixed point, it just needs to be a feature that you can identify in every frame. 

 

The purpose of drift correction/registration is to align that feature in every timepoint so that it *becomes* fixed. Then you can measure the motion of other things in the movie relative to that feature. I hope that makes sense.

 

Looking at the sample image ex.tif, I see a bright moving nuclear spot, but I don’t see any drift of the whole nucleus or cell. So I’m not sure what you want to correct for in that movie.

 

Best,

Theresa

 

 

On Oct 7, 2016, at 10:52 AM, Haitham Shaban <[hidden email]> wrote:

 

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Hi Theresa,


Thanks for your helpful suggestion.I think the main problem is that I do not have anything is fixed in my image stack. I have background around the nucleus that could be used as a fiducial, but since it is a part of the cell so it also has movement. Also it suppose that both of them have different motion behaviour so if we went to relative the motions for the spot based on the mask came out from the background, it will add artifact on the motion of the spot. Please correct me if I am wrong.


I found it is forbidden to attach images and videos within this group; please find download a small video for this link spot of my interest inside the nucleus, that is could be good example for real problem :)

https://www.sendbigfiles.com/76e07de679d73aaca015d6d13602e3a9/


Best regards,

Haitham

 

 

On Oct 7, 2016, at 11:58 AM, Julio Vazquez <[hidden email]> wrote:

 

Motion is only meaningful relative to a reference system. If you are tracking a single spot inside the nucleus, what is the reference system you are measuring the spot relative to? the slide may be moving relative to the camera; the cell may be moving relative to the slide; the nucleus may be moving relative to the cell... So I would say you first need to define precisely what motion you want to measure, and then choose and label a suitable reference object so that you can make your measurements. We faced a similar situation a few years ago when measuring chromosome motion, and our take was similar to that already mentioned of Wallace Marshall et al, namely to label two spots, and track their relative distance over time (Current Biology 11(16): 1227-1239. 2001). This gave us a measurement of the relative motion of two chromatin loci inside the nucleus, and was independent of the motion of everything else, including stage drift. If we assume that the two spots behave in a similar manner, then we can deduce the motion of each spot (or at  least some aspects of it). We were lucky  that the cells we were studying didn't have much motion themselves, so we could get a pretty good idea of the motion of each spot and verify that the two spots behaved in a very similar manner, but this is not always possible. Therefore, you need to take some time to think about the biological question you are interested in, which will lead you to a possible solution. For example, you could find the center of the nucleus, and use that as the reference (i.e. measure the distance between your spot and the center of the nucleus); or you could measure the shortest distance between your spot and the nuclear boundary; and so on.

 

------------------------------------

Theresa Swayne, Ph.D.

Manager

 

Herbert Irving Comprehensive Cancer Center

Columbia University Medical Center

1130 St. Nicholas Ave., Room 222A

New York, NY 10032

Phone: 212-851-4613

 


------------------------------------------------------------
This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
=================================