		Hi everyone,
 I am helping someone setup their system for live cell imaging with micro-manager. We observe a drift in northwest direction (order of 10 microns a second) in the XY stage, which is from Ludl and is controlled by MAC 5000 controller. Any tips on trouble-shooting? Seems the problem is with hardware. The OS (Windows 2000) complains about driver for MAC 2000 controller, but micro-manager controls the stage fine. However, I couldn't find the page to download the driver and test the XY stage with Ludl's software. Please point me to it if such a software is available. best Shalin mobile: +65-90694182 blog: shalin.wordpress.com Bioimaging Lab, Block-E3A, #7-10 Div of Bioengineering, NUS Singapore 117574 website: http://www.bioeng.nus.edu.sg/optbioimaging/colin/people.asp#shalinm | 
|   | 
		Start from the joystick first. I had experienced stage drift caused by a faulty joystick. Swap a joystick is much cheaper than a controller.  If the joystick is not the culprit, then, most likely the controller or the stage itself is faulty. Sometimes, the manufacture gives terminal test commands to test the controller via RS232, use HyperTerm to connect to the control interface directly. Check Ludl's site for such troubleshooting info. Good luck. Tao On Fri, Nov 20, 2009 at 12:01 AM, Shalin Mehta <[hidden email]> wrote: Hi everyone, -- http://tongtao.com | 
|   | 
				In reply to this post by Shalin Mehta
			 
		Shalin,
 Email the folks at Ludl. They are very helpful in resolving these kind of issues. I would also suspect the joystick though. There should be black push up/push down switch in the joystick. Make sure that it is completely in the off position. For one of our drift issues that was the culprit. -Prabhakar Shalin Mehta wrote: > Hi everyone, > > I am helping someone setup their system for live cell imaging with > micro-manager. We observe a drift in northwest direction (order of 10 > microns a second) in the XY stage, which is from Ludl and is > controlled by MAC 5000 controller. Any tips on trouble-shooting? Seems > the problem is with hardware. > The OS (Windows 2000) complains about driver for MAC 2000 controller, > but micro-manager controls the stage fine. However, I couldn't find > the page to download the driver and test the XY stage with Ludl's > software. Please point me to it if such a software is available. > > best > Shalin > > mobile: +65-90694182 > blog: shalin.wordpress.com > > Bioimaging Lab, Block-E3A, #7-10 > Div of Bioengineering, NUS Singapore 117574 > website: http://www.bioeng.nus.edu.sg/optbioimaging/colin/people.asp#shalinm > > > -- | 
|   | 
		Hi Shalin, 
 Me too, I have seen this in the past. You need to make sure the joystick springs back into the central position when deflected in any direction, if it doesn't look for those switches that Prabhakar refers to. You'll need to make sure that the joystick is in the central position when you turn on the Mac 5000. If the joystick is not centered then the system is initialized the stage will continue to drift in the direction the joystick was deflected. Cheers, Jason -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of B. Prabhakar Pandian Sent: 20 November 2009 12:13 To: [hidden email] Subject: Re: Drift in Ludl stage Shalin, Email the folks at Ludl. They are very helpful in resolving these kind of issues. I would also suspect the joystick though. There should be black push up/push down switch in the joystick. Make sure that it is completely in the off position. For one of our drift issues that was the culprit. -Prabhakar Shalin Mehta wrote: > Hi everyone, > > I am helping someone setup their system for live cell imaging with > micro-manager. We observe a drift in northwest direction (order of 10 > microns a second) in the XY stage, which is from Ludl and is > controlled by MAC 5000 controller. Any tips on trouble-shooting? Seems > the problem is with hardware. > The OS (Windows 2000) complains about driver for MAC 2000 controller, > but micro-manager controls the stage fine. However, I couldn't find > the page to download the driver and test the XY stage with Ludl's > software. Please point me to it if such a software is available. > > best > Shalin > > mobile: +65-90694182 > blog: shalin.wordpress.com > > Bioimaging Lab, Block-E3A, #7-10 > Div of Bioengineering, NUS Singapore 117574 > website: > > > -- | 
|   | 
				In reply to this post by Shalin Mehta
			 Hi Shalin, the problem might be much simpler - try to tighten the bolts on the stage. I noticed lateral drift after the microscope had been re-assembled after service. Vitaly From: Shalin Mehta <[hidden email]> To: [hidden email] Sent: Fri, November 20, 2009 12:01:27 AM Subject: Drift in Ludl stage Hi everyone, I am helping someone setup their system for live cell imaging with micro-manager. We observe a drift in northwest direction (order of 10 microns a second) in the XY stage, which is from Ludl and is controlled by MAC 5000 controller. Any tips on trouble-shooting? Seems the problem is with hardware. The OS (Windows 2000) complains about driver for MAC 2000 controller, but micro-manager controls the stage fine. However, I couldn't find the page to download the driver and test the XY stage with Ludl's software. Please point me to it if such a software is available. best Shalin mobile: +65-90694182 blog: shalin.wordpress.com Bioimaging Lab, Block-E3A, #7-10 Div of Bioengineering, NUS Singapore 117574 website: http://www.bioeng.nus.edu.sg/optbioimaging/colin/people.asp#shalinm | 
|    | Martin Hoppe-2 | 
|   | 
				In reply to this post by B. Prabhakar Pandian
			  *COMMERCIAL VENDOR INFORMATION*-------------------------------------------------------- Leica Microsystems is proud to introduce the newest addition to our line of super resolution confocal microscopes, the Leica TCS STED CW for diffraction unlimited research. Image the dynamics of living cells beyond the diffraction limit. STimulated Emission Depletion (STED), Continuous Wave (CW) technology is implemented purely with optics, providing instantaneous results. With the Leica TCS STED CW, you can utilize standard dyes and fluorescent proteins and see more detail than ever before. See it in San Diego, December 2nd through the 9th! If you will be in the San Diego, California area from December 2nd through the 9th and would like to have a demonstration, simply email: [hidden email] We will be in touch within two business days to schedule your appointment. If you are unable to make it to San Diego, join our Webcast December 4th. If you are unable to join us in San Diego, but would like to learn more about super resolution microscopy including STED and its application to biomedical research, please join us for a Leica sponsored webcast December 4th, 1 pm EST, 10 am PST, hosted by BioOptics World. The webcast will feature: Professor Stefan W. Hell, Director, Max Planck Institute for Biophysical Chemistry Professor Alberto Diaspro, University Professor of Applied Physics, Italian Institute of Technology Professor Stephan Sigrist, Professor in Genetics, Institute for Biology, Freie Universitaet Berlin To join the webcast, simply register at www.bioopticsworld.com You may also learn more about STED at our website, www.leica-microsystems.com With best regards, Martin Martin Hoppe, Ph.D. Director of Market Management & Confocal Application Management Life Science Division Leica Microsystems CMS GmbH Am Friedensplatz 3 | 68165 Mannheim (Germany) Phone : +49 621 7028 1100 | Fax : +49 621 7028 1180 Cell: +49 172 623 0409 Leica Microsystems CMS GmbH | GmbH mit Sitz in Wetzlar | Amtsgericht Wetzlar HRB 2432 Geschäftsführer : Dr. Stefan Traeger | Dr. David R. Martyr | Colin Davis | Dr. Wolf-Otto Reuter | 
|   | 
				In reply to this post by Vitaly Boyko
			 
		Thanks everyone who responded on and off the list. Joystick was the
 issue. I received documentation from Ludl that specifically advises to set the joystick in locked position (with those sliders) and not touching it for some time after the system is started. regards Shalin mobile: +65-90694182 blog: shalin.wordpress.com Bioimaging Lab, Block-E3A, #7-10 Div of Bioengineering, NUS Singapore 117574 website: http://www.bioeng.nus.edu.sg/optbioimaging/colin/people.asp#shalinm On Fri, Nov 20, 2009 at 9:52 PM, Vitaly Boyko <[hidden email]> wrote: > Hi Shalin, > > the problem might be much simpler - try to tighten the bolts on the stage. I > noticed lateral drift after the microscope had been re-assembled after > service. > > Vitaly > > ________________________________ > From: Shalin Mehta <[hidden email]> > To: [hidden email] > Sent: Fri, November 20, 2009 12:01:27 AM > Subject: Drift in Ludl stage > > Hi everyone, > > I am helping someone setup their system for live cell imaging with > micro-manager. We observe a drift in northwest direction (order of 10 > microns a second) in the XY stage, which is from Ludl and is > controlled by MAC 5000 controller. Any tips on trouble-shooting? Seems > the problem is with hardware. > The OS (Windows 2000) complains about driver for MAC 2000 controller, > but micro-manager controls the stage fine. However, I couldn't find > the page to download the driver and test the XY stage with Ludl's > software. Please point me to it if such a software is available. > > best > Shalin > > mobile: +65-90694182 > blog: shalin.wordpress.com > > Bioimaging Lab, Block-E3A, #7-10 > Div of Bioengineering, NUS Singapore 117574 > website: http://www.bioeng.nus.edu.sg/optbioimaging/colin/people.asp#shalinm > > | 