FRET results

classic Classic list List threaded Threaded
♦
♦
5 messages Options
ramesh kandpal ramesh kandpal
Reply | Threaded
Open this post in threaded view
♦
♦
|

FRET results

hi,
 
can anyone give some idea that which is the most appropriate laser wavelength for performing acceptor photobleaching method of CFP/YFP FRET. If 488nm laser is used for YFP excitation is there any possibility of loosing FRET data. What difference does it create  in FRET results if we use 514nm laser line.
 
thanks & best wishes,
 
RK
Periasamy, Ammasi (ap3t) Periasamy, Ammasi (ap3t)
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: FRET results

Usually we use 514 nm line from the argon laser to bleach or excite the YFP or Venus. 488nm is close to CFP excitation wavelength and you may bleach a lot of CFP molecule compared with 514nm.

 

 

Ammasi Periasamy, Ph.D.

Director, Keck Center for Cellular Imaging (KCCI)

Professor of Biology and Biomedical Engineering

Biology, Gilmer Hall (064), McCormick Rd

University of Virginia

Charlottesville, VA 22904

Voice: 434-243-7602 (Office); 982-4869 (lab)

Fax:434-982-5210; Email:[hidden email]

http//:www.kcci.virginia.edu

************************

Workshop on FRET Microscopy, March 9-13, 2010

http://www.kcci.virginia.edu/workshop/workshop2010/index.php

*************************

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of ramesh kandpal
Sent: Friday, April 03, 2009 7:56 AM
To: [hidden email]
Subject: FRET results

 

hi,

 

can anyone give some idea that which is the most appropriate laser wavelength for performing acceptor photobleaching method of CFP/YFP FRET. If 488nm laser is used for YFP excitation is there any possibility of loosing FRET data. What difference does it create  in FRET results if we use 514nm laser line.

 

thanks & best wishes,

 

RK

Zoltan Zoltan
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: FRET results

In reply to this post by ramesh kandpal
Hi,
 
We use 514 nm for yfp and 405 or 458 for cfp.  Please be aware of the potential artifact that comes from bleaching ypf and then exciting cfp with 405:
 
 

We do get a very strong (artificial) signal in the donor channel when using 405 nm even in  a control prep where there is no donor at all!  It seems to be a conversion of the yfp molecules to a form that can be easily excited by 405, but not nearly as much by 458. There was a paper about this around 2000 in Nature, I think. Some authors say they don't see this conversion, but we do.  Our recipe now is to excite cfp with 458.
  I hope this helps,
 
Zoltan
On Fri, Apr 3, 2009 at 12:55 PM, ramesh kandpal <[hidden email]> wrote:
hi,
 
can anyone give some idea that which is the most appropriate laser wavelength for performing acceptor photobleaching method of CFP/YFP FRET. If 488nm laser is used for YFP excitation is there any possibility of loosing FRET data. What difference does it create  in FRET results if we use 514nm laser line.
 
thanks & best wishes,
 
RK



--

Zoltan Cseresnyes
Facility manager, Imaging Suite
University of Cambridge, UK
Alberto Diaspro Alberto Diaspro
Reply | Threaded
Open this post in threaded view
♦
♦
|

Biophysics in Genoa -Summer 2009

Dear friends,

This is to remind you, on behalf of the EBSA Executive Committee, the Congress Organising Committee and the Italian Society for Pure and Applied Biophysics (SIBPA), to register for the VIIth European Biophysics Congress which will be held from 11th to 15th July 2009 in Genoa, Italy. Details of the Congress can be found on the Congress web site (http://www.ebsa2009.org) including the scientific programme, list of plenary speakers and registration details.

I am pleased to tell you that the deadline for abstract submission has been extended to 14th April 2009.

I am proud and  delighted to announce the following confirmed plenary lectures:
Special Plenary Lecture

Martin Chalfie,

2008 Nobel Laureate in Chemistry, Columbia University, USA.
Chair: Ranieri Bizzarri, NEST and Scuola Normale Superiore, Pisa, Italy.

Opening Lecture

Stefan W. Hell, Max-Planck-Institute for Biophysical Chemistry, Department of NanoBiophotonics, Göttingen, Germany.
Chair: Daniel Evanko, Nature Methods Editor.

Closing Lecture

Carlos Bustamante, Dept of Molecular and Cell Biology and of Physics, Howard Hughes Medical Institute, University of California, Berkeley, USA.
Chair: Bruno Samorì, University of Bologna, Italy.

Plenary Lectures (alphabetical order)

Philip Anfinrud, Laboratory of Chemical Physics, National Institutes of Health, Bethesda, MD, USA.
Fabio Beltram, NEST Laboratory, Scuola Normale Superiore, Pisa, Italy.
Edward H. Egelman, Dept. of Biochemistry and Molecular Genetics, University of Virginia, VA, USA.
Janos Hajdu, Laboratory of Structural Biology, Biomedical Centre, Uppsala University, Sweden.
Carol Robinson, Chemistry Department, University of Cambridge, UK.

Forty EBSA bursaries have now been awarded to young scientists from less favoured European countries. We are delighted to have received an enthusiastic response, and pleased that EBSA is able to contribute to European Biophysics in this positive way. As well, the Biophysical Society provided support for young scientists participation, and institutional and commercial sponsor allow to provide a number of benefits to all participants.

Congress costs have been kept to a minimum, and the registration includes the Congress dinner and all lunches for the duration of the congress. As 2009 is the 25th anniversary year of EBSA, the congress dinner will be an opportunity for special celebrations. We also arranged for excellent room prices at the Genova Hostel (see on the webpage) and also arranged free entry to museums on 16th July. Genoa is easily accessible by plane, train, coach and car. Inexpensive student accommodation is available as well as a full range of hotels. Genoa is a beautiful city on the Ligurian coast, an important port since before the Etruscans, with magnificent Renaissance architecture. The mountains, fishing harbours of Cinque Terre and beach resorts of the Italian Riviera will no doubt inspire you to extend your visit in July.
We are looking forward to seeing you in Genoa.

Yours sincerely

Alberto Diaspro
EBSA2009, Chairman
----------------------------------------------------
"Water slowly flowed under the sky" (Cesare Pavese)
----------------------------------------------------- 
Alberto Diaspro,
LAMBS IFOM IEO -MicroSCoBio, NBT-IIT, IBF-CNR
Department of Physics, University of Genoa, 
Via Dodecaneso 33, 16146 Genoa, Italy - 
fax +39-010314218 - tel +39 0103536426/309; 
URLs: www.lambs.it; 

Win in Science! ...link to http://www.ebsa2009.org
 ----------------------------------------------
 








Keith Morris Keith Morris
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: FRET results

In reply to this post by Periasamy, Ammasi (ap3t)

Yep we didn’t use the 488 line either, always the 514nm [the argon 514nm line is spot on for the YFP peak anyway]. Not sure is we even tried the 488nm line as it’s clearly not optimal for YFP [expect we did as it was there, and rapidly rejected it]. The *488nm line hits the tip of the CFP excitation peak [~4% excitation] and misses the YFP peak [excitation down from 100% at 514nm to 41% at 488nm], and so it is well worth avoiding the 488nm line and going for 514nm line for YFP if you can.

 

I know we did try the violet 405nm diode laser for CFP and it didn’t work as well as the ~450nm argon laser line [in fact the 405nm line was rather poor]. We were using a Leica SP2 with the optional FRET module, which made it all rather easier when FRET imaging. Most problems were getting FRET consistency between samples [worked well one month and no so good the next] rather than poor CFP/YFP FRET as such.

 

Regards

 

Keith

 

*I assume you know the excellent Invitrogen ‘spectral viewer’ for looking at fluorophore spectral excitation/emissions:

http://www.invitrogen.com/site/us/en/home/support/Research-Tools/Fluorescence-SpectraViewer.reg.uk.html

---------------------------------------------------------------------------
Dr Keith J. Morris,
Molecular Cytogenetics and Microscopy Core,
Laboratory 00/069 and 00/070,
The Wellcome Trust Centre for Human Genetics,
Roosevelt Drive,
Oxford  OX3 7BN,
United Kingdom.

Telephone:  +44 (0)1865 287568
Email:  [hidden email]
Web-pages: http://www.well.ox.ac.uk/cytogenetics/


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Periasamy, Ammasi (ap3t)
Sent: 03 April 2009 14:38
To: [hidden email]
Subject: Re: FRET results

 

Usually we use 514 nm line from the argon laser to bleach or excite the YFP or Venus. 488nm is close to CFP excitation wavelength and you may bleach a lot of CFP molecule compared with 514nm.

 

 

Ammasi Periasamy, Ph.D.

Director, Keck Center for Cellular Imaging (KCCI)

Professor of Biology and Biomedical Engineering

Biology, Gilmer Hall (064), McCormick Rd

University of Virginia

Charlottesville, VA 22904

Voice: 434-243-7602 (Office); 982-4869 (lab)

Fax:434-982-5210; Email:[hidden email]

http//:www.kcci.virginia.edu

************************

Workshop on FRET Microscopy, March 9-13, 2010

http://www.kcci.virginia.edu/workshop/workshop2010/index.php

*************************

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of ramesh kandpal
Sent: Friday, April 03, 2009 7:56 AM
To: [hidden email]
Subject: FRET results

 

hi,

 

can anyone give some idea that which is the most appropriate laser wavelength for performing acceptor photobleaching method of CFP/YFP FRET. If 488nm laser is used for YFP excitation is there any possibility of loosing FRET data. What difference does it create  in FRET results if we use 514nm laser line.

 

thanks & best wishes,

 

RK