To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Greetings to all The microscope facility is evaluating the purchase of an Olympus FV10i (confocal in a box). This equipment would be used for "standard" users (488, 543 and 633 nm) in order to reduce the work load on our Zeiss 510 Meta. I would appreciate all your opinions, and if possible, the price range to which you have been subjected (either at purchasing the equipment or as users at a facility). Please feel free to reply inter On or Off-list. Best regards Alejandro Roth PS Both oil-lens and water-immersion lens options are under evaluation. |
|
Arne Seitz |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Alejandro, the Olympus FV10i is a very nice instrument for routine work. Thus if you know that you are doing the same experiment the next years this is exactly the machine you want to buy. My experience in a facility is that you (your users) sometimes need more flexibility. And this is only partially offered by this microscope (in terms of optics but also/especially in terms of software). Therefore I would not consider this machine for a facility unless I know that I have a number of users which only need the capabilities of exactly that machine now and presumably also in the future. I'd prefer to buy a Leica SPE, Nikon C1 or Zeiss LSM700 for a facility. Just my 2c. Cheers Arne -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Alejandro Roth Sent: lundi 20 décembre 2010 16:05 To: [hidden email] Subject: FV10i opinions? (for a facility) ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Greetings to all The microscope facility is evaluating the purchase of an Olympus FV10i (confocal in a box). This equipment would be used for "standard" users (488, 543 and 633 nm) in order to reduce the work load on our Zeiss 510 Meta. I would appreciate all your opinions, and if possible, the price range to which you have been subjected (either at purchasing the equipment or as users at a facility). Please feel free to reply inter On or Off-list. Best regards Alejandro Roth PS Both oil-lens and water-immersion lens options are under evaluation. |
|
Sean Speese-2 |
|
|
In reply to this post by Alejandro Roth-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Alejandro, In line with the previous reply. We just purchased a Zeiss LSM700 for our lab and I have to say this is a really nice instrument and will be great for "standard" users in a core. I know price is always a consideration, but you might be surprised at what Zeiss offers this system for, so it is worth a look. No commercial interest Sean Speese Research Assistant Professor Oregon Health Sciences University Jungers Center for Neuroscience Research On Mon, Dec 20, 2010 at 7:04 AM, Alejandro Roth <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Greetings to all > > The microscope facility is evaluating the purchase of an Olympus FV10i > (confocal > in a box). This equipment would be used for "standard" users (488, 543 and > 633 > nm) in order to reduce the work load on our Zeiss 510 Meta. > I would appreciate all your opinions, and if possible, the price range to > which > you have been subjected (either at purchasing the equipment or as users at > a > facility). Please feel free to reply inter On or Off-list. > > Best regards > > Alejandro Roth > > PS > Both oil-lens and water-immersion lens options are under evaluation. > |
|
Josef Gotzmann |
|
|
In reply to this post by Alejandro Roth-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Alejandro, We have also tested the water-based FV10i-system and I totally agree with the other repliers. The major drawback of the setup is lack of optical flexibility (the optical zoom option cannot replace individual objective solutions). The system can be highly recommended, however, for confocal beginners. What I liked is the stability of the closed-box system for long-term experiments and its focus stability during multi-positioning. Moreover, the self-dispensing option raised no problems with water evaporation during 2 1/2 days of imaging, though at low-speed (definitely the system cannot be used for fast acquisition). What I missed was the option for plate screening and that the oil-based system also includes environmental control. In case the facility users do a lot of easy routine jobs you could buy one, although I am in line with Arne that for some 30-50k more you will get a more flexible system from Zeiss or Leica. Cheers Josef |
|
Weis, Michael-2 |
|
|
In reply to this post by Arne Seitz
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Seasons Greetings, Does anyone have any experience upgrading a 543 HeNe to 561 DPSS laser on a Leica SP2 AOBS confocal? I have inquired to Leica Canada and received the following. "In regards to the 561 laser for your microscope, we estimate that the laser will cost about $40K plus labor for the upgrade." This cost seems very high to me. Thanks, Michael Michael Weis Electron Microscopy & Digital Imaging Pacific Agri-Food Research Centre |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Mike, We are doing a similar upgrade... One additional 561 DPSS on SP5.. Initial quote was 76k USD :( In the end we bargained a bit, but still very expensive. Everything has to be shipped from Germany. Installation takes 4 working days. :(( Sigh..Good luck with upgrading.. Edna ________________________________________ From: Confocal Microscopy List [[hidden email]] On Behalf Of Weis, Michael [[hidden email]] Sent: Wednesday, December 22, 2010 1:43 AM To: [hidden email] Subject: 543 HeNe to 561 DPSS laser upgrade forLeica SP2 AOBS ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Seasons Greetings, Does anyone have any experience upgrading a 543 HeNe to 561 DPSS laser on a Leica SP2 AOBS confocal? I have inquired to Leica Canada and received the following. "In regards to the 561 laser for your microscope, we estimate that the laser will cost about $40K plus labor for the upgrade." This cost seems very high to me. Thanks, Michael Michael Weis Electron Microscopy & Digital Imaging Pacific Agri-Food Research Centre Best Regards, HU Xian,Edna Microscopy Core Manager, Research Centre of Excellence in Mechanobiology website: http://sites.google.com/a/blueprint.org/microscopycore/ Temasek Laboratories, Level 5 ,5A, Engineering Drive 1, Singapore 117411, DID:6601-1282, Email:[hidden email] |
|
Ralf Zenke |
|
|
In reply to this post by Weis, Michael-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Michael, The price for the 561nm upgrade seems to be correct - we paid approx. 24000 Euro two years ago (1 Euro ~ 1.3 Can.$), not including labor. But the advantage is a laser that excites commonly used red fluorophores with higher efficiency. Plus it has more power and does hopefully not age as fast as gas lasers. Nice regards, Ralf Ralf Zenke Max Planck Institute of Biochemistry Core Facility Am Klopferspitz 18 DE-82152 Martinsried near Munich GERMANY Phone: (+49) (89) 8578 3798 Fax: (+49) (89) 8578 2847 www.biochem.mpg.de Am Dienstag, den 21.12.2010, 12:43 -0500 schrieb Weis, Michael: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Seasons Greetings, > > Does anyone have any experience upgrading a 543 HeNe to 561 DPSS laser > on a Leica SP2 AOBS confocal? I have inquired to Leica Canada and > received the following. "In regards to the 561 laser for your > microscope, we estimate that the laser will cost about $40K plus labor > for the upgrade." This cost seems very high to me. > > Thanks, Michael > > Michael Weis > Electron Microscopy & Digital Imaging > Pacific Agri-Food Research Centre |
|
MICROSCOPIA CONFOCAL y CCD |
|
|
In reply to this post by Weis, Michael-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We upgraded 543 HeNe to 561 DPSS laser on a Leica SP2 AOBS confocal three years ago. It cost 30.000 . It works much better than 543 HeNe. It´s worth. Happy Christmas >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Seasons Greetings, > >Does anyone have any experience upgrading a 543 HeNe to 561 DPSS laser >on a Leica SP2 AOBS confocal? I have inquired to Leica Canada and >received the following. "In regards to the 561 laser for your >microscope, we estimate that the laser will cost about $40K plus labor >for the upgrade." This cost seems very high to me. > >Thanks, Michael > >Michael Weis >Electron Microscopy & Digital Imaging >Pacific Agri-Food Research Centre SERVICIO de MICROSCOPIA LASER CONFOCAL y MULTIDIMENSIONAL IN VIVO Mª Teresa Seisdedos Domínguez Oscar Hidalgo Blanco Roberto Encinar Calvo Begoña Pou Alonso [hidden email] Centro Investigaciones Biológicas (CIB) CSIC C/Ramiro de Maeztu, 9 28040 Madrid Phone: + 34-91 8373112 ext.4401 Fax: + 34-91 536 04 32 |
|
Tommaso Mello |
|
|
In reply to this post by Weis, Michael-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Micheal, we just ordered the 561 DPSS upgrade. It took quite some bargain to get around 24K (VAT not included). We are now waiting for the laser to be installed. Cheers, Tommaso On mar, 2010-12-21 at 12:43 -0500, Weis, Michael wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Seasons Greetings, > > Does anyone have any experience upgrading a 543 HeNe to 561 DPSS laser > on a Leica SP2 AOBS confocal? I have inquired to Leica Canada and > received the following. "In regards to the 561 laser for your > microscope, we estimate that the laser will cost about $40K plus labor > for the upgrade." This cost seems very high to me. > > Thanks, Michael > > Michael Weis > Electron Microscopy & Digital Imaging > Pacific Agri-Food Research Centre -- __________________ Tommaso Mello, PhD Gastroenterology Unit Dept. of Clinical Pathophysiology University of Florence Viale Pieraccini, 6 50139 - FIRENZE, ITALY Phone: +39.055.4271-294 FAX: +39.055.4271-297 |
|
|
In reply to this post by Weis, Michael-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Michael This is Chris Cathcart and I have started a new business to support the confocals in Canada. I have done a number of upgrades for the 543 to the 561 on the Nikon and would be pleased to take a look to see if I can help you out. I will be traveling thru Sunderland around the first of Feb can stop by to take a look Chris Cathcart Prismatic Solutions 919.714.1125 Can 647.975.5210 On Dec 21, 2010, at 12:43 PM, "Weis, Michael" <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Seasons Greetings, > > Does anyone have any experience upgrading a 543 HeNe to 561 DPSS laser > on a Leica SP2 AOBS confocal? I have inquired to Leica Canada and > received the following. "In regards to the 561 laser for your > microscope, we estimate that the laser will cost about $40K plus labor > for the upgrade." This cost seems very high to me. > > Thanks, Michael > > Michael Weis > Electron Microscopy & Digital Imaging > Pacific Agri-Food Research Centre |