FW: Yokogawa lenses

classic Classic list List threaded Threaded
♦
♦
18 messages Options
Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

FW: Yokogawa lenses

Yokogawa lenses
Comment from Tom Donelly.  Seems like UV plain won't work!

 

From: Donnelly, Tom [mailto:[hidden email]]
Sent: Saturday, 14 November 2009 1:35 PM
To: Guy Cox
Subject: Yokogawa lenses

Microlenses are plastic.

From day one Yo. Electric has been very clear m'lenses do not pass uv

Td
--------------------------
Sent using BlackBerry

This message was sent via Blackberry.



This email message, together with any attachments, is for the sole use of the intended recipient(s) and is the confidential information of Applied Precision Inc. If you are not the intended recipient, your review, use, disclosure, copying or dissemination of this email message or its attachments, or the information contained therein, is strictly prohibited. If you are not the intended recipient or if you think this email was sent to you in error, please notify the sender by reply email and delete this message and its attachments, as well as all copies, from your system.

No virus found in this incoming message.
Checked by AVG - www.avg.com
Version: 9.0.707 / Virus Database: 270.14.61/2497 - Release Date: 11/14/09 05:22:00

Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Yokogawa lenses

Yokogawa lenses
OK, from another commercial person (who does not want to be named) I've now heard that in fact the Yokogawa microlenses are in fact silica (as I had suspected).  But the snag is that there is another optic in the system which does not pass UV.  This is a bit more promising, since replacing the microlenses is not an option but replacing a transfer lens shouldn't be difficult.  Someone from Yokogawa must surely be reading this list.  One assumes, also, that they would be  interested in extending the application range of their system.  So, come on, Yokogawa, please give us the real story!
 
                                                                Guy
 

Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net

 


Alberto Diaspro Alberto Diaspro
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Yokogawa lenses

It is possible that they also have to take into account the possible  
heating due to intense UV
ciao
Alby




ISTITUTO ITALIANO
DI TECNOLOGIA

Prof. Alberto Diaspro
Scientific Head
Nanophysics
Via Morego, 30 16163 Genova
Tel: +39-010.71.781.503
Fax +39-010-72.03.21
Mobile +39-3666719968
www.iit.it


On Nov 14, 2009, at 12:55 PM, Guy Cox wrote:

> OK, from another commercial person (who does not want to be named)  
> I've now heard that in fact the Yokogawa microlenses are in fact  
> silica (as I had suspected).  But the snag is that there is another  
> optic in the system which does not pass UV.  This is a bit more  
> promising, since replacing the microlenses is not an option but  
> replacing a transfer lens shouldn't be difficult.  Someone from  
> Yokogawa must surely be reading this list.  One assumes, also, that  
> they would be  interested in extending the application range of  
> their system.  So, come on, Yokogawa, please give us the real story!
>
>                                                                 Guy
>
>
> Optical Imaging Techniques in Cell Biology
> by Guy Cox    CRC Press / Taylor & Francis
>     http://www.guycox.com/optical.htm
> ______________________________________________
> Associate Professor Guy Cox, MA, DPhil(Oxon)
> Electron Microscope Unit, Madsen Building F09,
> University of Sydney, NSW 2006
> ______________________________________________
> Phone +61 2 9351 3176     Fax +61 2 9351 7682
> Mobile 0413 281 861
> ______________________________________________
>      http://www.guycox.net
>
>
>

----------------------------------------
Alberto Diaspro
Head, Nanophysics Unit
Senior Scientist
The Italian Institute of Technology -IIT
Via Morego, 30
16163 - Genova (Italy)
phone: +39 010 71781503
mobile: +393666719968
fax:   +39 010 720321
http://www.iit.it
[hidden email]

Professor of Applied Physics
Department of Physics
University of Genova
Via Dodecaneso, 33
16146 Genova - Italy
tel.  +39 010 353 6426
fax. +39 010 314218
http://www.lambs.it
[hidden email]
-------------------------------------------------------
Renato A. Mortara Renato A. Mortara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Yokogawa lenses

Alby and Guy,

sure enough heating with UV is a concern.

I did wish someone from Yokogawa read this list and gave a conclusive  
answer whether any non UV-compatible element in the lightpath could be  
customized to be
UV friendly.

Grazie and Ciao,

Renato


Renato A. Mortara
Disciplina de Parasitologia
UNIFESP Escola Paulista de Medicina
R. Botucatu, 862 6o andar
04023-062
S?o Paulo SP
Brasil


Quoting Alberto Diaspro <[hidden email]>:

> It is possible that they also have to take into account the possible
> heating due to intense UV
> ciao
> Alby
>
>
>
>
> ISTITUTO ITALIANO
> DI TECNOLOGIA
>
> Prof. Alberto Diaspro
> Scientific Head
> Nanophysics
> Via Morego, 30 16163 Genova
> Tel: +39-010.71.781.503
> Fax +39-010-72.03.21
> Mobile +39-3666719968
> www.iit.it
>
>
> On Nov 14, 2009, at 12:55 PM, Guy Cox wrote:
>
>> OK, from another commercial person (who does not want to be named)  
>> I've now heard that in fact the Yokogawa microlenses are in fact  
>> silica (as I had suspected).  But the snag is that there is another  
>>  optic in the system which does not pass UV.  This is a bit more  
>> promising, since replacing the microlenses is not an option but  
>> replacing a transfer lens shouldn't be difficult.  Someone from  
>> Yokogawa must surely be reading this list.  One assumes, also, that  
>>  they would be  interested in extending the application range of  
>> their system.  So, come on, Yokogawa, please give us the real story!
>>
>>                                                                Guy
>>
>>
>> Optical Imaging Techniques in Cell Biology
>> by Guy Cox    CRC Press / Taylor & Francis
>>    http://www.guycox.com/optical.htm
>> ______________________________________________
>> Associate Professor Guy Cox, MA, DPhil(Oxon)
>> Electron Microscope Unit, Madsen Building F09,
>> University of Sydney, NSW 2006
>> ______________________________________________
>> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>> Mobile 0413 281 861
>> ______________________________________________
>>     http://www.guycox.net
>>
>>
>>
>
> ----------------------------------------
> Alberto Diaspro
> Head, Nanophysics Unit
> Senior Scientist
> The Italian Institute of Technology -IIT
> Via Morego, 30
> 16163 - Genova (Italy)
> phone: +39 010 71781503
> mobile: +393666719968
> fax:   +39 010 720321
> http://www.iit.it
> [hidden email]
>
> Professor of Applied Physics
> Department of Physics
> University of Genova
> Via Dodecaneso, 33
> 16146 Genova - Italy
> tel.  +39 010 353 6426
> fax. +39 010 314218
> http://www.lambs.it
> [hidden email]
> -------------------------------------------------------
Mizuho Shimizu Mizuho Shimizu
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Yokogawa lenses

Hi Alby, Guy, Renato and the list,

Please allow us to "enjoy" a weekend holiday before getting back with information to answer some of your questions regarding Yokogawa spinning disk.

I wish everybody to have a nice holiday, too!

Cheers,

Mizuho

***********************************************************    
Mizuho Shimizu
International Sales Team
BIO Group
B&A Center
Measurement Business Headquarters
Yokogawa Electric Corporation
2-3 Hokuyodai, Kanazawa-shi, Ishikawa
920-0177 JAPAN
Phone: +81-76-258-7028  Fax: +81-76-258-7029
 <http://www.yokogawa.com/scanner/index.htm>
***********************************************************
________________________________________
差出人: Confocal Microscopy List [[hidden email]] は Renato A. Mortara [[hidden email]] の代理
送信日時: 2009年11月15日 5:46
宛先: [hidden email]
件名: Re: Yokogawa lenses

Alby and Guy,

sure enough heating with UV is a concern.

I did wish someone from Yokogawa read this list and gave a conclusive
answer whether any non UV-compatible element in the lightpath could be
customized to be
UV friendly.

Grazie and Ciao,

Renato


Renato A. Mortara
Disciplina de Parasitologia
UNIFESP Escola Paulista de Medicina
R. Botucatu, 862 6o andar
04023-062
S?o Paulo SP
Brasil


Quoting Alberto Diaspro <[hidden email]>:

> It is possible that they also have to take into account the possible
> heating due to intense UV
> ciao
> Alby
>
>
>
>
> ISTITUTO ITALIANO
> DI TECNOLOGIA
>
> Prof. Alberto Diaspro
> Scientific Head
> Nanophysics
> Via Morego, 30 16163 Genova
> Tel: +39-010.71.781.503
> Fax +39-010-72.03.21
> Mobile +39-3666719968
> www.iit.it
>
>
> On Nov 14, 2009, at 12:55 PM, Guy Cox wrote:
>
>> OK, from another commercial person (who does not want to be named)
>> I've now heard that in fact the Yokogawa microlenses are in fact
>> silica (as I had suspected).  But the snag is that there is another
>>  optic in the system which does not pass UV.  This is a bit more
>> promising, since replacing the microlenses is not an option but
>> replacing a transfer lens shouldn't be difficult.  Someone from
>> Yokogawa must surely be reading this list.  One assumes, also, that
>>  they would be  interested in extending the application range of
>> their system.  So, come on, Yokogawa, please give us the real story!
>>
>>                                                                Guy
>>
>>
>> Optical Imaging Techniques in Cell Biology
>> by Guy Cox    CRC Press / Taylor & Francis
>>    http://www.guycox.com/optical.htm
>> ______________________________________________
>> Associate Professor Guy Cox, MA, DPhil(Oxon)
>> Electron Microscope Unit, Madsen Building F09,
>> University of Sydney, NSW 2006
>> ______________________________________________
>> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>> Mobile 0413 281 861
>> ______________________________________________
>>     http://www.guycox.net
>>
>>
>>
>
> ----------------------------------------
> Alberto Diaspro
> Head, Nanophysics Unit
> Senior Scientist
> The Italian Institute of Technology -IIT
> Via Morego, 30
> 16163 - Genova (Italy)
> phone: +39 010 71781503
> mobile: +393666719968
> fax:   +39 010 720321
> http://www.iit.it
> [hidden email]
>
> Professor of Applied Physics
> Department of Physics
> University of Genova
> Via Dodecaneso, 33
> 16146 Genova - Italy
> tel.  +39 010 353 6426
> fax. +39 010 314218
> http://www.lambs.it
> [hidden email]
> -------------------------------------------------------
Alberto Diaspro Alberto Diaspro
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Yokogawa lenses

In reply to this post by Renato A. Mortara
I am intersted in a precise answer...in case. Good luck.
Alby
On Nov 14, 2009, at 9:46 PM, Renato A. Mortara wrote:

> Alby and Guy,
>
> sure enough heating with UV is a concern.
>
> I did wish someone from Yokogawa read this list and gave a conclusive answer whether any non UV-compatible element in the lightpath could be customized to be
> UV friendly.
>
> Grazie and Ciao,
>
> Renato
>
>
> Renato A. Mortara
> Disciplina de Parasitologia
> UNIFESP Escola Paulista de Medicina
> R. Botucatu, 862 6o andar
> 04023-062
> S?o Paulo SP
> Brasil
>
>
> Quoting Alberto Diaspro <[hidden email]>:
>
>> It is possible that they also have to take into account the possible
>> heating due to intense UV
>> ciao
>> Alby
>>
>>
>>
>>
>> ISTITUTO ITALIANO
>> DI TECNOLOGIA
>>
>> Prof. Alberto Diaspro
>> Scientific Head
>> Nanophysics
>> Via Morego, 30 16163 Genova
>> Tel: +39-010.71.781.503
>> Fax +39-010-72.03.21
>> Mobile +39-3666719968
>> www.iit.it
>>
>>
>> On Nov 14, 2009, at 12:55 PM, Guy Cox wrote:
>>
>>> OK, from another commercial person (who does not want to be named)  I've now heard that in fact the Yokogawa microlenses are in fact  silica (as I had suspected).  But the snag is that there is another  optic in the system which does not pass UV.  This is a bit more  promising, since replacing the microlenses is not an option but  replacing a transfer lens shouldn't be difficult.  Someone from  Yokogawa must surely be reading this list.  One assumes, also, that  they would be  interested in extending the application range of  their system.  So, come on, Yokogawa, please give us the real story!
>>>
>>>                                                               Guy
>>>
>>>
>>> Optical Imaging Techniques in Cell Biology
>>> by Guy Cox    CRC Press / Taylor & Francis
>>>   http://www.guycox.com/optical.htm
>>> ______________________________________________
>>> Associate Professor Guy Cox, MA, DPhil(Oxon)
>>> Electron Microscope Unit, Madsen Building F09,
>>> University of Sydney, NSW 2006
>>> ______________________________________________
>>> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>>> Mobile 0413 281 861
>>> ______________________________________________
>>>    http://www.guycox.net
>>>
>>>
>>>
>>
>> ----------------------------------------
>> Alberto Diaspro
>> Head, Nanophysics Unit
>> Senior Scientist
>> The Italian Institute of Technology -IIT
>> Via Morego, 30
>> 16163 - Genova (Italy)
>> phone: +39 010 71781503
>> mobile: +393666719968
>> fax:   +39 010 720321
>> http://www.iit.it
>> [hidden email]
>>
>> Professor of Applied Physics
>> Department of Physics
>> University of Genova
>> Via Dodecaneso, 33
>> 16146 Genova - Italy
>> tel.  +39 010 353 6426
>> fax. +39 010 314218
>> http://www.lambs.it
>> [hidden email]
>> -------------------------------------------------------

----------------------------------------
Alberto Diaspro
Head, Nanophysics Unit
Senior Scientist
The Italian Institute of Technology -IIT
Via Morego, 30
16163 - Genova (Italy)
phone: +39 010 71781503
mobile: +393666719968
fax:   +39 010 720321
http://www.iit.it
[hidden email]

Professor of Applied Physics
Department of Physics
University of Genova
Via Dodecaneso, 33
16146 Genova - Italy
tel.  +39 010 353 6426
fax. +39 010 314218
http://www.lambs.it
[hidden email]
-------------------------------------------------------
Renato A. Mortara Renato A. Mortara
Reply | Threaded
Open this post in threaded view
♦
♦
|

RES: Yokogawa lenses

In reply to this post by Mizuho Shimizu
Mizuho
I hope you had a terrrific and relaxing weekend !

Cheers

Renato

-----Mensagem original-----
De: Confocal Microscopy List [mailto:[hidden email]] Em
nome de [hidden email]
Enviada em: sábado, 14 de novembro de 2009 21:14
Para: [hidden email]
Assunto: Re: Yokogawa lenses

Hi Alby, Guy, Renato and the list,

Please allow us to "enjoy" a weekend holiday before getting back with
information to answer some of your questions regarding Yokogawa spinning
disk.

I wish everybody to have a nice holiday, too!

Cheers,

Mizuho

***********************************************************    
Mizuho Shimizu
International Sales Team
BIO Group
B&A Center
Measurement Business Headquarters
Yokogawa Electric Corporation
2-3 Hokuyodai, Kanazawa-shi, Ishikawa
920-0177 JAPAN
Phone: +81-76-258-7028  Fax: +81-76-258-7029
<http://www.yokogawa.com/scanner/index.htm>
***********************************************************
________________________________________
差出人: Confocal Microscopy List [[hidden email]] は
Renato A. Mortara [[hidden email]] の代理
送信日�r: 2009年11月15日 5:46
宛先: [hidden email]
件名: Re: Yokogawa lenses

Alby and Guy,

sure enough heating with UV is a concern.

I did wish someone from Yokogawa read this list and gave a conclusive answer
whether any non UV-compatible element in the lightpath could be customized
to be UV friendly.

Grazie and Ciao,

Renato


Renato A. Mortara
Disciplina de Parasitologia
UNIFESP Escola Paulista de Medicina
R. Botucatu, 862 6o andar
04023-062
S?o Paulo SP
Brasil


Quoting Alberto Diaspro <[hidden email]>:

> It is possible that they also have to take into account the possible
> heating due to intense UV ciao Alby
>
>
>
>
> ISTITUTO ITALIANO
> DI TECNOLOGIA
>
> Prof. Alberto Diaspro
> Scientific Head
> Nanophysics
> Via Morego, 30 16163 Genova
> Tel: +39-010.71.781.503
> Fax +39-010-72.03.21
> Mobile +39-3666719968
> www.iit.it
>
>
> On Nov 14, 2009, at 12:55 PM, Guy Cox wrote:
>
>> OK, from another commercial person (who does not want to be named)
>> I've now heard that in fact the Yokogawa microlenses are in fact
>> silica (as I had suspected).  But the snag is that there is another  
>> optic in the system which does not pass UV.  This is a bit more
>> promising, since replacing the microlenses is not an option but
>> replacing a transfer lens shouldn't be difficult.  Someone from
>> Yokogawa must surely be reading this list.  One assumes, also, that  
>> they would be  interested in extending the application range of their
>> system.  So, come on, Yokogawa, please give us the real story!
>>
>>                                                                Guy
>>
>>
>> Optical Imaging Techniques in Cell Biology
>> by Guy Cox    CRC Press / Taylor & Francis
>>    http://www.guycox.com/optical.htm
>> ______________________________________________
>> Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope
>> Unit, Madsen Building F09, University of Sydney, NSW 2006
>> ______________________________________________
>> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>> Mobile 0413 281 861
>> ______________________________________________
>>     http://www.guycox.net
>>
>>
>>
>
> ----------------------------------------
> Alberto Diaspro
> Head, Nanophysics Unit
> Senior Scientist
> The Italian Institute of Technology -IIT Via Morego, 30
> 16163 - Genova (Italy)
> phone: +39 010 71781503
> mobile: +393666719968
> fax:   +39 010 720321
> http://www.iit.it
> [hidden email]
>
> Professor of Applied Physics
> Department of Physics
> University of Genova
> Via Dodecaneso, 33
> 16146 Genova - Italy
> tel.  +39 010 353 6426
> fax. +39 010 314218
> http://www.lambs.it
> [hidden email]
> -------------------------------------------------------
Renato A. Mortara Renato A. Mortara
Reply | Threaded
Open this post in threaded view
♦
♦
|

DAPI images and 3D recon with 405 line and Yokogawa head

In reply to this post by Alberto Diaspro
Good morning/evening everyone,

Hope all got a nice weekend !

From a potential customer point of view, what I would really like to see are
images/results with the 405 line to label DAPI for 3D reconstruction of
nuclei and other DNA-rich organelles using the Yokogawa head.

I am in the process of changing a "real" UV laser (Coherent 351/363) based -
old BioRad 1024 confocal and Zeiss optics (40x 1.2NA -UV and 100X 1.4
NON-UV) but from our point of view, even with the non UV-100X 1.4 we still
get reasonable (publishable) images of DAPI-stained structures. Since the UV
lines are quite powerful what we do is to just use 10x more DAPI and we are
in business with the 100X.

I would be grateful for any input yet again, especially published images of
DAPI with the 405 line using the Yokogawa CSU-X1 head.

Renato

Renato A. Mortara
Parasitology Division
UNIFESP - Escola Paulista de Medicina
Rua Botucatu, 862, 6th floor
São Paulo, SP
04023-062
Brazil
Phone: 55 11 5579-8306
Fax:     55 11 5571-1095
email: [hidden email]
home page: www.ecb.epm.br/~ramortara




-----Mensagem original-----
De: Confocal Microscopy List [mailto:[hidden email]] Em
nome de Alberto Diaspro
Enviada em: domingo, 15 de novembro de 2009 14:34
Para: [hidden email]
Assunto: Re: Yokogawa lenses

I am intersted in a precise answer...in case. Good luck.
Alby
On Nov 14, 2009, at 9:46 PM, Renato A. Mortara wrote:

> Alby and Guy,
>
> sure enough heating with UV is a concern.
>
> I did wish someone from Yokogawa read this list and gave a conclusive
> answer whether any non UV-compatible element in the lightpath could be
customized to be UV friendly.

>
> Grazie and Ciao,
>
> Renato
>
>
> Renato A. Mortara
> Disciplina de Parasitologia
> UNIFESP Escola Paulista de Medicina
> R. Botucatu, 862 6o andar
> 04023-062
> S?o Paulo SP
> Brasil
>
>
> Quoting Alberto Diaspro <[hidden email]>:
>
>> It is possible that they also have to take into account the possible
>> heating due to intense UV ciao Alby
>>
>>
>>
>>
>> ISTITUTO ITALIANO
>> DI TECNOLOGIA
>>
>> Prof. Alberto Diaspro
>> Scientific Head
>> Nanophysics
>> Via Morego, 30 16163 Genova
>> Tel: +39-010.71.781.503
>> Fax +39-010-72.03.21
>> Mobile +39-3666719968
>> www.iit.it
>>
>>
>> On Nov 14, 2009, at 12:55 PM, Guy Cox wrote:
>>
>>> OK, from another commercial person (who does not want to be named)  I've
now heard that in fact the Yokogawa microlenses are in fact  silica (as I
had suspected).  But the snag is that there is another  optic in the system
which does not pass UV.  This is a bit more  promising, since replacing the
microlenses is not an option but  replacing a transfer lens shouldn't be
difficult.  Someone from  Yokogawa must surely be reading this list.  One
assumes, also, that  they would be  interested in extending the application
range of  their system.  So, come on, Yokogawa, please give us the real
story!
>>>
>>>                                                               Guy
>>>
>>>
Cameron, Lisa Cameron, Lisa
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: DAPI images and 3D recon with 405 line and Yokogawa head

Hi Renato,

My users have been taking DAPI images on my CSU-x1 system from Andor
Technologies with a 405nm laser with good results. You can see: Ganem NJ,
Godinho SA, Pellman D. A Mechanism Linking Extra Centrosomes to Chromosomal
Instability. Nature (2009) July 9;460:278-82. I know Neil Ganem did
deconvolution to help identify merotelic kinetochores but may have ended up
publishing the max intensity projections and single planes. If you want to see
the raw images of the DAPI illuminated with 405nm, I can ask the author. We have
many other users who also collect DAPI or Pacific Blue images with 405nm laser
illumination, but are not doing deconvolution. Please feel free to contact me
for more information.

- Lisa


---------------------------------------
Lisa Cameron, Ph.D.
Director of Confocal and Light Microscopy
Dana Farber Cancer Institute
44 Binney St.; JF 621
Boston, MA 02115
Office phone: 617-582-8824
Fax: 617-582-8750
[hidden email]
http://microscopy.dfci.harvard.edu


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Renato Mortara
Sent: Monday, November 16, 2009 5:35 AM
To: [hidden email]
Subject: [CONFOCALMICROSCOPY] DAPI images and 3D recon with 405 line and
Yokogawa head

Good morning/evening everyone,

Hope all got a nice weekend !

From a potential customer point of view, what I would really like to see are
images/results with the 405 line to label DAPI for 3D reconstruction of
nuclei and other DNA-rich organelles using the Yokogawa head.

I am in the process of changing a "real" UV laser (Coherent 351/363) based -
old BioRad 1024 confocal and Zeiss optics (40x 1.2NA -UV and 100X 1.4
NON-UV) but from our point of view, even with the non UV-100X 1.4 we still
get reasonable (publishable) images of DAPI-stained structures. Since the UV
lines are quite powerful what we do is to just use 10x more DAPI and we are
in business with the 100X.

I would be grateful for any input yet again, especially published images of
DAPI with the 405 line using the Yokogawa CSU-X1 head.

Renato

Renato A. Mortara
Parasitology Division
UNIFESP - Escola Paulista de Medicina
Rua Botucatu, 862, 6th floor
São Paulo, SP
04023-062
Brazil
Phone: 55 11 5579-8306
Fax:     55 11 5571-1095
email: [hidden email]
home page: www.ecb.epm.br/~ramortara



The information in this e-mail is intended only for the person to whom it is
addressed. If you believe this e-mail was sent to you in error and the e-mail
contains patient information, please contact the Partners Compliance HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you in error
but does not contain patient information, please contact the sender and properly
dispose of the e-mail.
Mizuho Shimizu Mizuho Shimizu
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Yokogawa lenses

In reply to this post by Renato A. Mortara
Dear Renato, and the list.

Yes, I enjoyed gorgeous autumn weather over the weekend!


While I am not in a position to officially represent Yokogawa, as Renato has mentioned already, official comment from Yokogawa, if requested, will be to ask to use the CSU spinning disk confocal system within 400nm-700nm wavelength range for best performance with a good reason.

As Lisa kindly advised, it is quite OK to use the CSUX1 system at 405 nm.
So, if your are talking about 405nm, it is within our specification range.

However, if you are thinking about using  351/363 nm, as is well understood, imaging at UV range below 400nm could be very difficult, which is not unique to the imaging with the CSU system but universal to all optical and/or imaging devices.
We have well confirmed the performance of the CSU confocal system including all components such as the microscope, laser and fibers, all filters, detector, etc.at 400-700 nm range.
However, we have no conclusive data on the use below 400nm range, in reality.

Our engineers don't think there could be any simple and easy way to optimize the CSU head for imaging at UV range, since every piece of optical parts both inside and outside of the CSU head requires optimization for UV.
It may be possible to capture images below 400nm, but, we cannot guarantee good-enough light efficiency, confocality or chromatic aberration correction, and thus cannot endorse it.

I am sorry for not being able to offer any quick solution but please kindly understand there is no magic in good UV imaging.


Mizuho Shimizu

Yokogawa Electric Corporation




-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Renato Mortara
Sent: Monday, November 16, 2009 7:06 PM
To: [hidden email]
Subject: RES: Yokogawa lenses

Mizuho
I hope you had a terrrific and relaxing weekend !

Cheers

Renato

-----Mensagem original-----
De: Confocal Microscopy List [mailto:[hidden email]] Em nome de [hidden email] Enviada em: sábado, 14 de novembro de 2009 21:14
Para: [hidden email]
Assunto: Re: Yokogawa lenses

Hi Alby, Guy, Renato and the list,

Please allow us to "enjoy" a weekend holiday before getting back with information to answer some of your questions regarding Yokogawa spinning disk.

I wish everybody to have a nice holiday, too!

Cheers,

Mizuho

***********************************************************    
Mizuho Shimizu
International Sales Team
BIO Group
B&A Center
Measurement Business Headquarters
Yokogawa Electric Corporation
2-3 Hokuyodai, Kanazawa-shi, Ishikawa
920-0177 JAPAN
Phone: +81-76-258-7028  Fax: +81-76-258-7029 <http://www.yokogawa.com/scanner/index.htm>
***********************************************************
________________________________________
差出人: Confocal Microscopy List [[hidden email]] は
Renato A. Mortara [[hidden email]] の代理
送信日�r: 2009年11月15日 5:46
宛先: [hidden email]
件名: Re: Yokogawa lenses

Alby and Guy,

sure enough heating with UV is a concern.

I did wish someone from Yokogawa read this list and gave a conclusive answer whether any non UV-compatible element in the lightpath could be customized to be UV friendly.

Grazie and Ciao,

Renato


Renato A. Mortara
Disciplina de Parasitologia
UNIFESP Escola Paulista de Medicina
R. Botucatu, 862 6o andar
04023-062
S?o Paulo SP
Brasil


Quoting Alberto Diaspro <[hidden email]>:

> It is possible that they also have to take into account the possible
> heating due to intense UV ciao Alby
>
>
>
>
> ISTITUTO ITALIANO
> DI TECNOLOGIA
>
> Prof. Alberto Diaspro
> Scientific Head
> Nanophysics
> Via Morego, 30 16163 Genova
> Tel: +39-010.71.781.503
> Fax +39-010-72.03.21
> Mobile +39-3666719968
> www.iit.it
>
>
> On Nov 14, 2009, at 12:55 PM, Guy Cox wrote:
>
>> OK, from another commercial person (who does not want to be named)
>> I've now heard that in fact the Yokogawa microlenses are in fact
>> silica (as I had suspected).  But the snag is that there is another
>> optic in the system which does not pass UV.  This is a bit more
>> promising, since replacing the microlenses is not an option but
>> replacing a transfer lens shouldn't be difficult.  Someone from
>> Yokogawa must surely be reading this list.  One assumes, also, that
>> they would be  interested in extending the application range of their
>> system.  So, come on, Yokogawa, please give us the real story!
>>
>>                                                                Guy
>>
>>
>> Optical Imaging Techniques in Cell Biology
>> by Guy Cox    CRC Press / Taylor & Francis
>>    http://www.guycox.com/optical.htm
>> ______________________________________________
>> Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope
>> Unit, Madsen Building F09, University of Sydney, NSW 2006
>> ______________________________________________
>> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>> Mobile 0413 281 861
>> ______________________________________________
>>     http://www.guycox.net
>>
>>
>>
>
> ----------------------------------------
> Alberto Diaspro
> Head, Nanophysics Unit
> Senior Scientist
> The Italian Institute of Technology -IIT Via Morego, 30
> 16163 - Genova (Italy)
> phone: +39 010 71781503
> mobile: +393666719968
> fax:   +39 010 720321
> http://www.iit.it
> [hidden email]
>
> Professor of Applied Physics
> Department of Physics
> University of Genova
> Via Dodecaneso, 33
> 16146 Genova - Italy
> tel.  +39 010 353 6426
> fax. +39 010 314218
> http://www.lambs.it
> [hidden email]
> -------------------------------------------------------
Renato A. Mortara Renato A. Mortara
Reply | Threaded
Open this post in threaded view
♦
♦
|

RES: Yokogawa lenses

Dear Mizuho
Thanks for the input anyway.
Best,

Renato

-----Mensagem original-----
De: Confocal Microscopy List [mailto:[hidden email]] Em
nome de [hidden email]
Enviada em: segunda-feira, 16 de novembro de 2009 23:20
Para: [hidden email]
Assunto: Re: Yokogawa lenses

Dear Renato, and the list.

Yes, I enjoyed gorgeous autumn weather over the weekend!


While I am not in a position to officially represent Yokogawa, as Renato has
mentioned already, official comment from Yokogawa, if requested, will be to
ask to use the CSU spinning disk confocal system within 400nm-700nm
wavelength range for best performance with a good reason.

As Lisa kindly advised, it is quite OK to use the CSUX1 system at 405 nm.
So, if your are talking about 405nm, it is within our specification range.

However, if you are thinking about using  351/363 nm, as is well understood,
imaging at UV range below 400nm could be very difficult, which is not unique
to the imaging with the CSU system but universal to all optical and/or
imaging devices.
We have well confirmed the performance of the CSU confocal system including
all components such as the microscope, laser and fibers, all filters,
detector, etc.at 400-700 nm range.
However, we have no conclusive data on the use below 400nm range, in
reality.

Our engineers don't think there could be any simple and easy way to optimize
the CSU head for imaging at UV range, since every piece of optical parts
both inside and outside of the CSU head requires optimization for UV.
It may be possible to capture images below 400nm, but, we cannot guarantee
good-enough light efficiency, confocality or chromatic aberration
correction, and thus cannot endorse it.

I am sorry for not being able to offer any quick solution but please kindly
understand there is no magic in good UV imaging.


Mizuho Shimizu

Yokogawa Electric Corporation




-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Renato Mortara
Sent: Monday, November 16, 2009 7:06 PM
To: [hidden email]
Subject: RES: Yokogawa lenses

Mizuho
I hope you had a terrrific and relaxing weekend !

Cheers

Renato

-----Mensagem original-----
De: Confocal Microscopy List [mailto:[hidden email]] Em
nome de [hidden email] Enviada em: sábado, 14 de novembro
de 2009 21:14
Para: [hidden email]
Assunto: Re: Yokogawa lenses

Hi Alby, Guy, Renato and the list,

Please allow us to "enjoy" a weekend holiday before getting back with
information to answer some of your questions regarding Yokogawa spinning
disk.

I wish everybody to have a nice holiday, too!

Cheers,

Mizuho

***********************************************************    
Mizuho Shimizu
International Sales Team
BIO Group
B&A Center
Measurement Business Headquarters
Yokogawa Electric Corporation
2-3 Hokuyodai, Kanazawa-shi, Ishikawa
920-0177 JAPAN
Phone: +81-76-258-7028  Fax: +81-76-258-7029
<http://www.yokogawa.com/scanner/index.htm>
***********************************************************
________________________________________
差出人: Confocal Microscopy List [[hidden email]] は
Renato A. Mortara [[hidden email]] の代理
送信日�r: 2009年11月15日 5:46
宛先: [hidden email]
件名: Re: Yokogawa lenses

Alby and Guy,

sure enough heating with UV is a concern.

I did wish someone from Yokogawa read this list and gave a conclusive answer
whether any non UV-compatible element in the lightpath could be customized
to be UV friendly.

Grazie and Ciao,

Renato


Renato A. Mortara
Disciplina de Parasitologia
UNIFESP Escola Paulista de Medicina
R. Botucatu, 862 6o andar
04023-062
S?o Paulo SP
Brasil


Quoting Alberto Diaspro <[hidden email]>:

> It is possible that they also have to take into account the possible
> heating due to intense UV ciao Alby
>
>
>
>
> ISTITUTO ITALIANO
> DI TECNOLOGIA
>
> Prof. Alberto Diaspro
> Scientific Head
> Nanophysics
> Via Morego, 30 16163 Genova
> Tel: +39-010.71.781.503
> Fax +39-010-72.03.21
> Mobile +39-3666719968
> www.iit.it
>
>
> On Nov 14, 2009, at 12:55 PM, Guy Cox wrote:
>
>> OK, from another commercial person (who does not want to be named)
>> I've now heard that in fact the Yokogawa microlenses are in fact
>> silica (as I had suspected).  But the snag is that there is another
>> optic in the system which does not pass UV.  This is a bit more
>> promising, since replacing the microlenses is not an option but
>> replacing a transfer lens shouldn't be difficult.  Someone from
>> Yokogawa must surely be reading this list.  One assumes, also, that
>> they would be  interested in extending the application range of their
>> system.  So, come on, Yokogawa, please give us the real story!
>>
>>                                                                Guy
>>
>>
>> Optical Imaging Techniques in Cell Biology
>> by Guy Cox    CRC Press / Taylor & Francis
>>    http://www.guycox.com/optical.htm
>> ______________________________________________
>> Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope
>> Unit, Madsen Building F09, University of Sydney, NSW 2006
>> ______________________________________________
>> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>> Mobile 0413 281 861
>> ______________________________________________
>>     http://www.guycox.net
>>
>>
>>
>
> ----------------------------------------
> Alberto Diaspro
> Head, Nanophysics Unit
> Senior Scientist
> The Italian Institute of Technology -IIT Via Morego, 30
> 16163 - Genova (Italy)
> phone: +39 010 71781503
> mobile: +393666719968
> fax:   +39 010 720321
> http://www.iit.it
> [hidden email]
>
> Professor of Applied Physics
> Department of Physics
> University of Genova
> Via Dodecaneso, 33
> 16146 Genova - Italy
> tel.  +39 010 353 6426
> fax. +39 010 314218
> http://www.lambs.it
> [hidden email]
> -------------------------------------------------------
Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Yokogawa lenses

In reply to this post by Mizuho Shimizu
Well, I'm glad your weekend was good (so was mine actually, with kangaroos all around and lyre-birds playing in the garden).

But your considered opinion after all that was to give us NO information.  We are scientists, we are a community.  The relationship between manufacturers and users is what drives advances in the microscopy field.  Please be a player in this - it really is to your advantage.   What we need to know is simple - something you can tell us in an instant.  What is the wavelength transmission range of your micro-lenses?  For what wavelength are they chromatically corrected?  What other components in your head might affect the wavelength transmission?

These are not great secrets of huge commercial value.  In fact, someone will certainly make the relevant measurements and post them if you don't.  But in the end collaboration is what drives progress, so I hope you will change your mind and decide to become part of the community.


                                                                            Guy Cox



Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net
-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email]
Sent: Tuesday, 17 November 2009 12:20 PM
To: [hidden email]
Subject: Re: Yokogawa lenses

Dear Renato, and the list.

Yes, I enjoyed gorgeous autumn weather over the weekend!


While I am not in a position to officially represent Yokogawa, as Renato has mentioned already, official comment from Yokogawa, if requested, will be to ask to use the CSU spinning disk confocal system within 400nm-700nm wavelength range for best performance with a good reason.

As Lisa kindly advised, it is quite OK to use the CSUX1 system at 405 nm.
So, if your are talking about 405nm, it is within our specification range.

However, if you are thinking about using  351/363 nm, as is well understood, imaging at UV range below 400nm could be very difficult, which is not unique to the imaging with the CSU system but universal to all optical and/or imaging devices.
We have well confirmed the performance of the CSU confocal system including all components such as the microscope, laser and fibers, all filters, detector, etc.at 400-700 nm range.
However, we have no conclusive data on the use below 400nm range, in reality.

Our engineers don't think there could be any simple and easy way to optimize the CSU head for imaging at UV range, since every piece of optical parts both inside and outside of the CSU head requires optimization for UV.
It may be possible to capture images below 400nm, but, we cannot guarantee good-enough light efficiency, confocality or chromatic aberration correction, and thus cannot endorse it.

I am sorry for not being able to offer any quick solution but please kindly understand there is no magic in good UV imaging.


Mizuho Shimizu

Yokogawa Electric Corporation




-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Renato Mortara
Sent: Monday, November 16, 2009 7:06 PM
To: [hidden email]
Subject: RES: Yokogawa lenses

Mizuho
I hope you had a terrrific and relaxing weekend !

Cheers

Renato

-----Mensagem original-----
De: Confocal Microscopy List [mailto:[hidden email]] Em nome de [hidden email] Enviada em: sábado, 14 de novembro de 2009 21:14
Para: [hidden email]
Assunto: Re: Yokogawa lenses

Hi Alby, Guy, Renato and the list,

Please allow us to "enjoy" a weekend holiday before getting back with information to answer some of your questions regarding Yokogawa spinning disk.

I wish everybody to have a nice holiday, too!

Cheers,

Mizuho

***********************************************************    
Mizuho Shimizu
International Sales Team
BIO Group
B&A Center
Measurement Business Headquarters
Yokogawa Electric Corporation
2-3 Hokuyodai, Kanazawa-shi, Ishikawa
920-0177 JAPAN
Phone: +81-76-258-7028  Fax: +81-76-258-7029 <http://www.yokogawa.com/scanner/index.htm>
***********************************************************
________________________________________
差出人: Confocal Microscopy List [[hidden email]] は
Renato A. Mortara [[hidden email]] の代理
送信日�r: 2009年11月15日 5:46
宛先: [hidden email]
件名: Re: Yokogawa lenses

Alby and Guy,

sure enough heating with UV is a concern.

I did wish someone from Yokogawa read this list and gave a conclusive answer whether any non UV-compatible element in the lightpath could be customized to be UV friendly.

Grazie and Ciao,

Renato


Renato A. Mortara
Disciplina de Parasitologia
UNIFESP Escola Paulista de Medicina
R. Botucatu, 862 6o andar
04023-062
S?o Paulo SP
Brasil


Quoting Alberto Diaspro <[hidden email]>:

> It is possible that they also have to take into account the possible
> heating due to intense UV ciao Alby
>
>
>
>
> ISTITUTO ITALIANO
> DI TECNOLOGIA
>
> Prof. Alberto Diaspro
> Scientific Head
> Nanophysics
> Via Morego, 30 16163 Genova
> Tel: +39-010.71.781.503
> Fax +39-010-72.03.21
> Mobile +39-3666719968
> www.iit.it
>
>
> On Nov 14, 2009, at 12:55 PM, Guy Cox wrote:
>
>> OK, from another commercial person (who does not want to be named)
>> I've now heard that in fact the Yokogawa microlenses are in fact
>> silica (as I had suspected).  But the snag is that there is another
>> optic in the system which does not pass UV.  This is a bit more
>> promising, since replacing the microlenses is not an option but
>> replacing a transfer lens shouldn't be difficult.  Someone from
>> Yokogawa must surely be reading this list.  One assumes, also, that
>> they would be  interested in extending the application range of their
>> system.  So, come on, Yokogawa, please give us the real story!
>>
>>                                                                Guy
>>
>>
>> Optical Imaging Techniques in Cell Biology
>> by Guy Cox    CRC Press / Taylor & Francis
>>    http://www.guycox.com/optical.htm
>> ______________________________________________
>> Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope
>> Unit, Madsen Building F09, University of Sydney, NSW 2006
>> ______________________________________________
>> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>> Mobile 0413 281 861
>> ______________________________________________
>>     http://www.guycox.net
>>
>>
>>
>
> ----------------------------------------
> Alberto Diaspro
> Head, Nanophysics Unit
> Senior Scientist
> The Italian Institute of Technology -IIT Via Morego, 30
> 16163 - Genova (Italy)
> phone: +39 010 71781503
> mobile: +393666719968
> fax:   +39 010 720321
> http://www.iit.it
> [hidden email]
>
> Professor of Applied Physics
> Department of Physics
> University of Genova
> Via Dodecaneso, 33
> 16146 Genova - Italy
> tel.  +39 010 353 6426
> fax. +39 010 314218
> http://www.lambs.it
> [hidden email]
> -------------------------------------------------------

No virus found in this incoming message.
Checked by AVG - www.avg.com
Version: 9.0.707 / Virus Database: 270.14.61/2497 - Release Date: 11/17/09 06:53:00
Neeraj Gohad-2 Neeraj Gohad-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Yokogawa lenses

I agree with Guy, there is much to be accomplished, the companies should stopped treating scientist as just end users and we should stop treating them as just vendors. Both parties will definitely benefit if there is real dialogue.  

Best,

Neeraj.


Neeraj V. Gohad, Ph.D.
Postdoctoral Fellow
Okeanos Research Group
Department of Biological Sciences
132 Long Hall
Clemson University

Please note my new email address: [hidden email]




-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Guy Cox
Sent: Tuesday, November 17, 2009 6:18 AM
To: [hidden email]
Subject: Re: Yokogawa lenses

Well, I'm glad your weekend was good (so was mine actually, with kangaroos all around and lyre-birds playing in the garden).

But your considered opinion after all that was to give us NO information.  We are scientists, we are a community.  The relationship between manufacturers and users is what drives advances in the microscopy field.  Please be a player in this - it really is to your advantage.   What we need to know is simple - something you can tell us in an instant.  What is the wavelength transmission range of your micro-lenses?  For what wavelength are they chromatically corrected?  What other components in your head might affect the wavelength transmission?

These are not great secrets of huge commercial value.  In fact, someone will certainly make the relevant measurements and post them if you don't.  But in the end collaboration is what drives progress, so I hope you will change your mind and decide to become part of the community.


                                                                            Guy Cox



Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net
-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email]
Sent: Tuesday, 17 November 2009 12:20 PM
To: [hidden email]
Subject: Re: Yokogawa lenses

Dear Renato, and the list.

Yes, I enjoyed gorgeous autumn weather over the weekend!


While I am not in a position to officially represent Yokogawa, as Renato has mentioned already, official comment from Yokogawa, if requested, will be to ask to use the CSU spinning disk confocal system within 400nm-700nm wavelength range for best performance with a good reason.

As Lisa kindly advised, it is quite OK to use the CSUX1 system at 405 nm.
So, if your are talking about 405nm, it is within our specification range.

However, if you are thinking about using  351/363 nm, as is well understood, imaging at UV range below 400nm could be very difficult, which is not unique to the imaging with the CSU system but universal to all optical and/or imaging devices.
We have well confirmed the performance of the CSU confocal system including all components such as the microscope, laser and fibers, all filters, detector, etc.at 400-700 nm range.
However, we have no conclusive data on the use below 400nm range, in reality.

Our engineers don't think there could be any simple and easy way to optimize the CSU head for imaging at UV range, since every piece of optical parts both inside and outside of the CSU head requires optimization for UV.
It may be possible to capture images below 400nm, but, we cannot guarantee good-enough light efficiency, confocality or chromatic aberration correction, and thus cannot endorse it.

I am sorry for not being able to offer any quick solution but please kindly understand there is no magic in good UV imaging.


Mizuho Shimizu

Yokogawa Electric Corporation




-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Renato Mortara
Sent: Monday, November 16, 2009 7:06 PM
To: [hidden email]
Subject: RES: Yokogawa lenses

Mizuho
I hope you had a terrrific and relaxing weekend !

Cheers

Renato

-----Mensagem original-----
De: Confocal Microscopy List [mailto:[hidden email]] Em nome de [hidden email] Enviada em: sábado, 14 de novembro de 2009 21:14
Para: [hidden email]
Assunto: Re: Yokogawa lenses

Hi Alby, Guy, Renato and the list,

Please allow us to "enjoy" a weekend holiday before getting back with information to answer some of your questions regarding Yokogawa spinning disk.

I wish everybody to have a nice holiday, too!

Cheers,

Mizuho

***********************************************************    
Mizuho Shimizu
International Sales Team
BIO Group
B&A Center
Measurement Business Headquarters
Yokogawa Electric Corporation
2-3 Hokuyodai, Kanazawa-shi, Ishikawa
920-0177 JAPAN
Phone: +81-76-258-7028  Fax: +81-76-258-7029 <http://www.yokogawa.com/scanner/index.htm>
***********************************************************
________________________________________
差出人: Confocal Microscopy List [[hidden email]] は
Renato A. Mortara [[hidden email]] の代理
送信日�r: 2009年11月15日 5:46
宛先: [hidden email]
件名: Re: Yokogawa lenses

Alby and Guy,

sure enough heating with UV is a concern.

I did wish someone from Yokogawa read this list and gave a conclusive answer whether any non UV-compatible element in the lightpath could be customized to be UV friendly.

Grazie and Ciao,

Renato


Renato A. Mortara
Disciplina de Parasitologia
UNIFESP Escola Paulista de Medicina
R. Botucatu, 862 6o andar
04023-062
S?o Paulo SP
Brasil


Quoting Alberto Diaspro <[hidden email]>:

> It is possible that they also have to take into account the possible
> heating due to intense UV ciao Alby
>
>
>
>
> ISTITUTO ITALIANO
> DI TECNOLOGIA
>
> Prof. Alberto Diaspro
> Scientific Head
> Nanophysics
> Via Morego, 30 16163 Genova
> Tel: +39-010.71.781.503
> Fax +39-010-72.03.21
> Mobile +39-3666719968
> www.iit.it
>
>
> On Nov 14, 2009, at 12:55 PM, Guy Cox wrote:
>
>> OK, from another commercial person (who does not want to be named)
>> I've now heard that in fact the Yokogawa microlenses are in fact
>> silica (as I had suspected).  But the snag is that there is another
>> optic in the system which does not pass UV.  This is a bit more
>> promising, since replacing the microlenses is not an option but
>> replacing a transfer lens shouldn't be difficult.  Someone from
>> Yokogawa must surely be reading this list.  One assumes, also, that
>> they would be  interested in extending the application range of their
>> system.  So, come on, Yokogawa, please give us the real story!
>>
>>                                                                Guy
>>
>>
>> Optical Imaging Techniques in Cell Biology
>> by Guy Cox    CRC Press / Taylor & Francis
>>    http://www.guycox.com/optical.htm
>> ______________________________________________
>> Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope
>> Unit, Madsen Building F09, University of Sydney, NSW 2006
>> ______________________________________________
>> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>> Mobile 0413 281 861
>> ______________________________________________
>>     http://www.guycox.net
>>
>>
>>
>
> ----------------------------------------
> Alberto Diaspro
> Head, Nanophysics Unit
> Senior Scientist
> The Italian Institute of Technology -IIT Via Morego, 30
> 16163 - Genova (Italy)
> phone: +39 010 71781503
> mobile: +393666719968
> fax:   +39 010 720321
> http://www.iit.it
> [hidden email]
>
> Professor of Applied Physics
> Department of Physics
> University of Genova
> Via Dodecaneso, 33
> 16146 Genova - Italy
> tel.  +39 010 353 6426
> fax. +39 010 314218
> http://www.lambs.it
> [hidden email]
> -------------------------------------------------------

No virus found in this incoming message.
Checked by AVG - www.avg.com
Version: 9.0.707 / Virus Database: 270.14.61/2497 - Release Date: 11/17/09 06:53:00
Paul Maddox Paul Maddox
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Yokogawa lenses

Part of the "problem" here might be in way the question is being asked.  In
my experience, the "end users" should take the initiative to experiment and
try new applications of existing equipment.  Thereby we can empirically test
these ideas on real world experimental applications.  However, to properly
execute this, we need to know certain things such as if UV light (eg
300-400nm) will damage the system.  The question of if it will work or not
is up to us to determine (ultimately by peer review).  The proper question
to Yokogawa in this case is will we damage the system in doing so?

Clearly, the greater the exchange of information the better for all
involved.  But in practical terms, there is never a 100% free flow of info
and I doubt there will be in the very near future.

These are my two cents (Canadian, so ±1.8 US)...

Paul

Paul S. Maddox, PhD
Assistant Professor
Institute for Research in Immunology and Cancer
Dept of Pathology and Cell Biol, U. de Montreal
P.O. Box 6128, Station Centre-Ville
Montréal QC  H3C 3J7
CANADA

Courier:
2900, boulevard Édouard-Montpetit
Pavillon Marcelle-Coutu, Quai 20
Montreal QC  H3T 1J4
CANADA

[hidden email]
Ph: 514-343-7894
Fax: 514-343-6843




On 17/11/09 11:17 AM, "Neeraj Gohad" <[hidden email]> wrote:

> I agree with Guy, there is much to be accomplished, the companies should
> stopped treating scientist as just end users and we should stop treating them
> as just vendors. Both parties will definitely benefit if there is real
> dialogue.  
>
> Best,
>
> Neeraj.
>
>
> Neeraj V. Gohad, Ph.D.
> Postdoctoral Fellow
> Okeanos Research Group
> Department of Biological Sciences
> 132 Long Hall
> Clemson University
>
> Please note my new email address: [hidden email]
>
>
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On
> Behalf Of Guy Cox
> Sent: Tuesday, November 17, 2009 6:18 AM
> To: [hidden email]
> Subject: Re: Yokogawa lenses
>
> Well, I'm glad your weekend was good (so was mine actually, with kangaroos all
> around and lyre-birds playing in the garden).
>
> But your considered opinion after all that was to give us NO information.  We
> are scientists, we are a community.  The relationship between manufacturers
> and users is what drives advances in the microscopy field.  Please be a player
> in this - it really is to your advantage.   What we need to know is simple -
> something you can tell us in an instant.  What is the wavelength transmission
> range of your micro-lenses?  For what wavelength are they chromatically
> corrected?  What other components in your head might affect the wavelength
> transmission?
>
> These are not great secrets of huge commercial value.  In fact, someone will
> certainly make the relevant measurements and post them if you don't.  But in
> the end collaboration is what drives progress, so I hope you will change your
> mind and decide to become part of the community.
>
>
>                  
> Guy Cox
>
>
>
> Optical Imaging Techniques in Cell Biology
> by Guy Cox    CRC Press / Taylor & Francis
>     http://www.guycox.com/optical.htm
> ______________________________________________
> Associate Professor Guy Cox, MA, DPhil(Oxon)
> Electron Microscope Unit, Madsen Building F09,
> University of Sydney, NSW 2006
> ______________________________________________
> Phone +61 2 9351 3176     Fax +61 2 9351 7682
> Mobile 0413 281 861
> ______________________________________________
>      http://www.guycox.net
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On
> Behalf Of [hidden email]
> Sent: Tuesday, 17 November 2009 12:20 PM
> To: [hidden email]
> Subject: Re: Yokogawa lenses
>
> Dear Renato, and the list.
>
> Yes, I enjoyed gorgeous autumn weather over the weekend!
>
>
> While I am not in a position to officially represent Yokogawa, as Renato has
> mentioned already, official comment from Yokogawa, if requested, will be to
> ask to use the CSU spinning disk confocal system within 400nm-700nm wavelength
> range for best performance with a good reason.
>
> As Lisa kindly advised, it is quite OK to use the CSUX1 system at 405 nm.
> So, if your are talking about 405nm, it is within our specification range.
>
> However, if you are thinking about using  351/363 nm, as is well understood,
> imaging at UV range below 400nm could be very difficult, which is not unique
> to the imaging with the CSU system but universal to all optical and/or imaging
> devices.
> We have well confirmed the performance of the CSU confocal system including
> all components such as the microscope, laser and fibers, all filters,
> detector, etc.at 400-700 nm range.
> However, we have no conclusive data on the use below 400nm range, in reality.
>
> Our engineers don't think there could be any simple and easy way to optimize
> the CSU head for imaging at UV range, since every piece of optical parts both
> inside and outside of the CSU head requires optimization for UV.
> It may be possible to capture images below 400nm, but, we cannot guarantee
> good-enough light efficiency, confocality or chromatic aberration correction,
> and thus cannot endorse it.
>
> I am sorry for not being able to offer any quick solution but please kindly
> understand there is no magic in good UV imaging.
>
>
> Mizuho Shimizu
>
> Yokogawa Electric Corporation
>
>
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On
> Behalf Of Renato Mortara
> Sent: Monday, November 16, 2009 7:06 PM
> To: [hidden email]
> Subject: RES: Yokogawa lenses
>
> Mizuho
> I hope you had a terrrific and relaxing weekend !
>
> Cheers
>
> Renato
>
> -----Mensagem original-----
> De: Confocal Microscopy List [mailto:[hidden email]] Em nome
> de [hidden email] Enviada em: sábado, 14 de novembro de 2009
> 21:14
> Para: [hidden email]
> Assunto: Re: Yokogawa lenses
>
> Hi Alby, Guy, Renato and the list,
>
> Please allow us to "enjoy" a weekend holiday before getting back with
> information to answer some of your questions regarding Yokogawa spinning disk.
>
> I wish everybody to have a nice holiday, too!
>
> Cheers,
>
> Mizuho
>
> ***********************************************************
> Mizuho Shimizu
> International Sales Team
> BIO Group
> B&A Center
> Measurement Business Headquarters
> Yokogawa Electric Corporation
> 2-3 Hokuyodai, Kanazawa-shi, Ishikawa
> 920-0177 JAPAN
> Phone: +81-76-258-7028  Fax: +81-76-258-7029
> <http://www.yokogawa.com/scanner/index.htm>
> ***********************************************************
> ________________________________________
> 差出人: Confocal Microscopy List [[hidden email]] は
> Renato A. Mortara [[hidden email]] の代理
> 送信日時: 2009年11月15日 5:46
> 宛先: [hidden email]
> 件名: Re: Yokogawa lenses
>
> Alby and Guy,
>
> sure enough heating with UV is a concern.
>
> I did wish someone from Yokogawa read this list and gave a conclusive answer
> whether any non UV-compatible element in the lightpath could be customized to
> be UV friendly.
>
> Grazie and Ciao,
>
> Renato
>
>
> Renato A. Mortara
> Disciplina de Parasitologia
> UNIFESP Escola Paulista de Medicina
> R. Botucatu, 862 6o andar
> 04023-062
> S?o Paulo SP
> Brasil
>
>
> Quoting Alberto Diaspro <[hidden email]>:
>
>> It is possible that they also have to take into account the possible
>> heating due to intense UV ciao Alby
>>
>>
>>
>>
>> ISTITUTO ITALIANO
>> DI TECNOLOGIA
>>
>> Prof. Alberto Diaspro
>> Scientific Head
>> Nanophysics
>> Via Morego, 30 16163 Genova
>> Tel: +39-010.71.781.503
>> Fax +39-010-72.03.21
>> Mobile +39-3666719968
>> www.iit.it
>>
>>
>> On Nov 14, 2009, at 12:55 PM, Guy Cox wrote:
>>
>>> OK, from another commercial person (who does not want to be named)
>>> I've now heard that in fact the Yokogawa microlenses are in fact
>>> silica (as I had suspected).  But the snag is that there is another
>>> optic in the system which does not pass UV.  This is a bit more
>>> promising, since replacing the microlenses is not an option but
>>> replacing a transfer lens shouldn't be difficult.  Someone from
>>> Yokogawa must surely be reading this list.  One assumes, also, that
>>> they would be  interested in extending the application range of their
>>> system.  So, come on, Yokogawa, please give us the real story!
>>>
>>>                                                                Guy
>>>
>>>
>>> Optical Imaging Techniques in Cell Biology
>>> by Guy Cox    CRC Press / Taylor & Francis
>>>    http://www.guycox.com/optical.htm
>>> ______________________________________________
>>> Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope
>>> Unit, Madsen Building F09, University of Sydney, NSW 2006
>>> ______________________________________________
>>> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>>> Mobile 0413 281 861
>>> ______________________________________________
>>>     http://www.guycox.net
>>>
>>>
>>>
>>
>> ----------------------------------------
>> Alberto Diaspro
>> Head, Nanophysics Unit
>> Senior Scientist
>> The Italian Institute of Technology -IIT Via Morego, 30
>> 16163 - Genova (Italy)
>> phone: +39 010 71781503
>> mobile: +393666719968
>> fax:   +39 010 720321
>> http://www.iit.it
>> [hidden email]
>>
>> Professor of Applied Physics
>> Department of Physics
>> University of Genova
>> Via Dodecaneso, 33
>> 16146 Genova - Italy
>> tel.  +39 010 353 6426
>> fax. +39 010 314218
>> http://www.lambs.it
>> [hidden email]
>> -------------------------------------------------------
>
> No virus found in this incoming message.
> Checked by AVG - www.avg.com
> Version: 9.0.707 / Virus Database: 270.14.61/2497 - Release Date: 11/17/09
> 06:53:00
Renato A. Mortara Renato A. Mortara
Reply | Threaded
Open this post in threaded view
♦
♦
|

RES: Yokogawa lenses

What is great in this list is that all the information flows along opinions and ideas. I agree with Guy as well as Paul. Unfortunately, I do not have enough $$ to play around with the systems,fibers,etc, just to try to implement an application... - my 2 cents in Brazilian Reais, ± $0.10

Best

Renato

-----Mensagem original-----
De: Confocal Microscopy List [mailto:[hidden email]] Em nome de Paul Maddox
Enviada em: terça-feira, 17 de novembro de 2009 14:54
Para: [hidden email]
Assunto: Re: Yokogawa lenses

Part of the "problem" here might be in way the question is being asked.  In my experience, the "end users" should take the initiative to experiment and try new applications of existing equipment.  Thereby we can empirically test these ideas on real world experimental applications.  However, to properly execute this, we need to know certain things such as if UV light (eg
300-400nm) will damage the system.  The question of if it will work or not is up to us to determine (ultimately by peer review).  The proper question to Yokogawa in this case is will we damage the system in doing so?

Clearly, the greater the exchange of information the better for all involved.  But in practical terms, there is never a 100% free flow of info and I doubt there will be in the very near future.

These are my two cents (Canadian, so ±1.8 US)...

Paul

Paul S. Maddox, PhD
Assistant Professor
Institute for Research in Immunology and Cancer Dept of Pathology and Cell Biol, U. de Montreal P.O. Box 6128, Station Centre-Ville Montréal QC  H3C 3J7 CANADA

Courier:
2900, boulevard Édouard-Montpetit
Pavillon Marcelle-Coutu, Quai 20
Montreal QC  H3T 1J4
CANADA

[hidden email]
Ph: 514-343-7894
Fax: 514-343-6843




On 17/11/09 11:17 AM, "Neeraj Gohad" <[hidden email]> wrote:

> I agree with Guy, there is much to be accomplished, the companies should
> stopped treating scientist as just end users and we should stop treating them
> as just vendors. Both parties will definitely benefit if there is real
> dialogue.  
>
> Best,
>
> Neeraj.
>
>
> Neeraj V. Gohad, Ph.D.
> Postdoctoral Fellow
> Okeanos Research Group
> Department of Biological Sciences
> 132 Long Hall
> Clemson University
>
> Please note my new email address: [hidden email]
>
>
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On
> Behalf Of Guy Cox
> Sent: Tuesday, November 17, 2009 6:18 AM
> To: [hidden email]
> Subject: Re: Yokogawa lenses
>
> Well, I'm glad your weekend was good (so was mine actually, with kangaroos all
> around and lyre-birds playing in the garden).
>
> But your considered opinion after all that was to give us NO information.  We
> are scientists, we are a community.  The relationship between manufacturers
> and users is what drives advances in the microscopy field.  Please be a player
> in this - it really is to your advantage.   What we need to know is simple -
> something you can tell us in an instant.  What is the wavelength transmission
> range of your micro-lenses?  For what wavelength are they chromatically
> corrected?  What other components in your head might affect the wavelength
> transmission?
>
> These are not great secrets of huge commercial value.  In fact, someone will
> certainly make the relevant measurements and post them if you don't.  But in
> the end collaboration is what drives progress, so I hope you will change your
> mind and decide to become part of the community.
>
>
>                  
> Guy Cox
>
>
>
> Optical Imaging Techniques in Cell Biology
> by Guy Cox    CRC Press / Taylor & Francis
>     http://www.guycox.com/optical.htm
> ______________________________________________
> Associate Professor Guy Cox, MA, DPhil(Oxon)
> Electron Microscope Unit, Madsen Building F09,
> University of Sydney, NSW 2006
> ______________________________________________
> Phone +61 2 9351 3176     Fax +61 2 9351 7682
> Mobile 0413 281 861
> ______________________________________________
>      http://www.guycox.net
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On
> Behalf Of [hidden email]
> Sent: Tuesday, 17 November 2009 12:20 PM
> To: [hidden email]
> Subject: Re: Yokogawa lenses
>
> Dear Renato, and the list.
>
> Yes, I enjoyed gorgeous autumn weather over the weekend!
>
>
> While I am not in a position to officially represent Yokogawa, as Renato has
> mentioned already, official comment from Yokogawa, if requested, will be to
> ask to use the CSU spinning disk confocal system within 400nm-700nm wavelength
> range for best performance with a good reason.
>
> As Lisa kindly advised, it is quite OK to use the CSUX1 system at 405 nm.
> So, if your are talking about 405nm, it is within our specification range.
>
> However, if you are thinking about using  351/363 nm, as is well understood,
> imaging at UV range below 400nm could be very difficult, which is not unique
> to the imaging with the CSU system but universal to all optical and/or imaging
> devices.
> We have well confirmed the performance of the CSU confocal system including
> all components such as the microscope, laser and fibers, all filters,
> detector, etc.at 400-700 nm range.
> However, we have no conclusive data on the use below 400nm range, in reality.
>
> Our engineers don't think there could be any simple and easy way to optimize
> the CSU head for imaging at UV range, since every piece of optical parts both
> inside and outside of the CSU head requires optimization for UV.
> It may be possible to capture images below 400nm, but, we cannot guarantee
> good-enough light efficiency, confocality or chromatic aberration correction,
> and thus cannot endorse it.
>
> I am sorry for not being able to offer any quick solution but please kindly
> understand there is no magic in good UV imaging.
>
>
> Mizuho Shimizu
>
> Yokogawa Electric Corporation
>
>
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On
> Behalf Of Renato Mortara
> Sent: Monday, November 16, 2009 7:06 PM
> To: [hidden email]
> Subject: RES: Yokogawa lenses
>
> Mizuho
> I hope you had a terrrific and relaxing weekend !
>
> Cheers
>
> Renato
>
> -----Mensagem original-----
> De: Confocal Microscopy List [mailto:[hidden email]] Em nome
> de [hidden email] Enviada em: sábado, 14 de novembro de 2009
> 21:14
> Para: [hidden email]
> Assunto: Re: Yokogawa lenses
>
> Hi Alby, Guy, Renato and the list,
>
> Please allow us to "enjoy" a weekend holiday before getting back with
> information to answer some of your questions regarding Yokogawa spinning disk.
>
> I wish everybody to have a nice holiday, too!
>
> Cheers,
>
> Mizuho
>
> ***********************************************************
> Mizuho Shimizu
> International Sales Team
> BIO Group
> B&A Center
> Measurement Business Headquarters
> Yokogawa Electric Corporation
> 2-3 Hokuyodai, Kanazawa-shi, Ishikawa
> 920-0177 JAPAN
> Phone: +81-76-258-7028  Fax: +81-76-258-7029
> <http://www.yokogawa.com/scanner/index.htm>
> ***********************************************************
> ________________________________________
> 差出人: Confocal Microscopy List [[hidden email]] は
> Renato A. Mortara [[hidden email]] の代理
> 送信日時: 2009年11月15日 5:46
> 宛先: [hidden email]
> 件名: Re: Yokogawa lenses
>
> Alby and Guy,
>
> sure enough heating with UV is a concern.
>
> I did wish someone from Yokogawa read this list and gave a conclusive answer
> whether any non UV-compatible element in the lightpath could be customized to
> be UV friendly.
>
> Grazie and Ciao,
>
> Renato
>
>
> Renato A. Mortara
> Disciplina de Parasitologia
> UNIFESP Escola Paulista de Medicina
> R. Botucatu, 862 6o andar
> 04023-062
> S?o Paulo SP
> Brasil
>
>
> Quoting Alberto Diaspro <[hidden email]>:
>
>> It is possible that they also have to take into account the possible
>> heating due to intense UV ciao Alby
>>
>>
>>
>>
>> ISTITUTO ITALIANO
>> DI TECNOLOGIA
>>
>> Prof. Alberto Diaspro
>> Scientific Head
>> Nanophysics
>> Via Morego, 30 16163 Genova
>> Tel: +39-010.71.781.503
>> Fax +39-010-72.03.21
>> Mobile +39-3666719968
>> www.iit.it
>>
>>
>> On Nov 14, 2009, at 12:55 PM, Guy Cox wrote:
>>
>>> OK, from another commercial person (who does not want to be named)
>>> I've now heard that in fact the Yokogawa microlenses are in fact
>>> silica (as I had suspected).  But the snag is that there is another
>>> optic in the system which does not pass UV.  This is a bit more
>>> promising, since replacing the microlenses is not an option but
>>> replacing a transfer lens shouldn't be difficult.  Someone from
>>> Yokogawa must surely be reading this list.  One assumes, also, that
>>> they would be  interested in extending the application range of their
>>> system.  So, come on, Yokogawa, please give us the real story!
>>>
>>>                                                                Guy
>>>
>>>
>>> Optical Imaging Techniques in Cell Biology
>>> by Guy Cox    CRC Press / Taylor & Francis
>>>    http://www.guycox.com/optical.htm
>>> ______________________________________________
>>> Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope
>>> Unit, Madsen Building F09, University of Sydney, NSW 2006
>>> ______________________________________________
>>> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>>> Mobile 0413 281 861
>>> ______________________________________________
>>>     http://www.guycox.net
>>>
>>>
>>>
>>
>> ----------------------------------------
>> Alberto Diaspro
>> Head, Nanophysics Unit
>> Senior Scientist
>> The Italian Institute of Technology -IIT Via Morego, 30
>> 16163 - Genova (Italy)
>> phone: +39 010 71781503
>> mobile: +393666719968
>> fax:   +39 010 720321
>> http://www.iit.it
>> [hidden email]
>>
>> Professor of Applied Physics
>> Department of Physics
>> University of Genova
>> Via Dodecaneso, 33
>> 16146 Genova - Italy
>> tel.  +39 010 353 6426
>> fax. +39 010 314218
>> http://www.lambs.it
>> [hidden email]
>> -------------------------------------------------------
>
> No virus found in this incoming message.
> Checked by AVG - www.avg.com
> Version: 9.0.707 / Virus Database: 270.14.61/2497 - Release Date: 11/17/09
> 06:53:00
Richard Berman Richard Berman
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: RES: Yokogawa lenses

******** commercial interest *************

Hi Renato and Confocal List: As I stated in a previous post, we will be
investigating the potential to operate the CSU at 375nm shortly. We
expect the laser in about 2 weeks. As has been mentioned by another
member, the potential to damage optics must always be considered. We do
not believe that damage due to heating will be a factor since the heat
is distributed over a fairly large area of the spinning disk. In fact if
there is an issue it will be with the pinhole disk where the power
densities are much higher. Damage due to UV is a greater concern but can
be assessed. Many higher quality lenslet arrays are made from fused
silica.  Transmission tests will enable us to determine if either of the
disks are a polymer or both are fused silica. If both disks are silica
they are probably not UV grade fused silica, but since we are not going
to hard UV the standard grade will almost certainly show no degradation
at 375nm. And by the way, single mode fibers that are rated to work at
375nm are made from the same fused silica as fibers for longer
wavelengths. The power densities in the fibers are orders of magnitude
higher than what will be hitting the disks and so degradation will be
evident in the fiber long before the disks. This is something we are
going to be assessing more carefully moving forward and should not be
considered a guarantee of performance at this time.

And on a slightly different note, we have been very successful modifying
the CSU to use lasers above 700nm, with emission wavelengths approaching
800nm. We have given the modifications a commercial name, BOREALIS. We
are very excited about this development and our partner Quorum
Technologies (also a commercial interest) has successfully installed two
such systems with great results. Just to say that moving beyond the
standard 400nm - 700nm range is possible.

Best Regards, Richard

Renato Mortara wrote:

> What is great in this list is that all the information flows along opinions and ideas. I agree with Guy as well as Paul. Unfortunately, I do not have enough $$ to play around with the systems,fibers,etc, just to try to implement an application... - my 2 cents in Brazilian Reais, ± $0.10
>
> Best
>
> Renato
>
> -----Mensagem original-----
> De: Confocal Microscopy List [mailto:[hidden email]] Em nome de Paul Maddox
> Enviada em: terça-feira, 17 de novembro de 2009 14:54
> Para: [hidden email]
> Assunto: Re: Yokogawa lenses
>
> Part of the "problem" here might be in way the question is being asked.  In my experience, the "end users" should take the initiative to experiment and try new applications of existing equipment.  Thereby we can empirically test these ideas on real world experimental applications.  However, to properly execute this, we need to know certain things such as if UV light (eg
> 300-400nm) will damage the system.  The question of if it will work or not is up to us to determine (ultimately by peer review).  The proper question to Yokogawa in this case is will we damage the system in doing so?
>
> Clearly, the greater the exchange of information the better for all involved.  But in practical terms, there is never a 100% free flow of info and I doubt there will be in the very near future.
>
> These are my two cents (Canadian, so ±1.8 US)...
>
> Paul
>
> Paul S. Maddox, PhD
> Assistant Professor
> Institute for Research in Immunology and Cancer Dept of Pathology and Cell Biol, U. de Montreal P.O. Box 6128, Station Centre-Ville Montréal QC  H3C 3J7 CANADA
>
> Courier:
> 2900, boulevard Édouard-Montpetit
> Pavillon Marcelle-Coutu, Quai 20
> Montreal QC  H3T 1J4
> CANADA
>
> [hidden email]
> Ph: 514-343-7894
> Fax: 514-343-6843
>
>
>
>
> On 17/11/09 11:17 AM, "Neeraj Gohad" <[hidden email]> wrote:
>
>  
>> I agree with Guy, there is much to be accomplished, the companies should
>> stopped treating scientist as just end users and we should stop treating them
>> as just vendors. Both parties will definitely benefit if there is real
>> dialogue.  
>>
>> Best,
>>
>> Neeraj.
>>
>>
>> Neeraj V. Gohad, Ph.D.
>> Postdoctoral Fellow
>> Okeanos Research Group
>> Department of Biological Sciences
>> 132 Long Hall
>> Clemson University
>>
>> Please note my new email address: [hidden email]
>>
>>
>>
>>
>> -----Original Message-----
>> From: Confocal Microscopy List [mailto:[hidden email]] On
>> Behalf Of Guy Cox
>> Sent: Tuesday, November 17, 2009 6:18 AM
>> To: [hidden email]
>> Subject: Re: Yokogawa lenses
>>
>> Well, I'm glad your weekend was good (so was mine actually, with kangaroos all
>> around and lyre-birds playing in the garden).
>>
>> But your considered opinion after all that was to give us NO information.  We
>> are scientists, we are a community.  The relationship between manufacturers
>> and users is what drives advances in the microscopy field.  Please be a player
>> in this - it really is to your advantage.   What we need to know is simple -
>> something you can tell us in an instant.  What is the wavelength transmission
>> range of your micro-lenses?  For what wavelength are they chromatically
>> corrected?  What other components in your head might affect the wavelength
>> transmission?
>>
>> These are not great secrets of huge commercial value.  In fact, someone will
>> certainly make the relevant measurements and post them if you don't.  But in
>> the end collaboration is what drives progress, so I hope you will change your
>> mind and decide to become part of the community.
>>
>>
>>                  
>> Guy Cox
>>
>>
>>
>> Optical Imaging Techniques in Cell Biology
>> by Guy Cox    CRC Press / Taylor & Francis
>>     http://www.guycox.com/optical.htm
>> ______________________________________________
>> Associate Professor Guy Cox, MA, DPhil(Oxon)
>> Electron Microscope Unit, Madsen Building F09,
>> University of Sydney, NSW 2006
>> ______________________________________________
>> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>> Mobile 0413 281 861
>> ______________________________________________
>>      http://www.guycox.net
>> -----Original Message-----
>> From: Confocal Microscopy List [mailto:[hidden email]] On
>> Behalf Of [hidden email]
>> Sent: Tuesday, 17 November 2009 12:20 PM
>> To: [hidden email]
>> Subject: Re: Yokogawa lenses
>>
>> Dear Renato, and the list.
>>
>> Yes, I enjoyed gorgeous autumn weather over the weekend!
>>
>>
>> While I am not in a position to officially represent Yokogawa, as Renato has
>> mentioned already, official comment from Yokogawa, if requested, will be to
>> ask to use the CSU spinning disk confocal system within 400nm-700nm wavelength
>> range for best performance with a good reason.
>>
>> As Lisa kindly advised, it is quite OK to use the CSUX1 system at 405 nm.
>> So, if your are talking about 405nm, it is within our specification range.
>>
>> However, if you are thinking about using  351/363 nm, as is well understood,
>> imaging at UV range below 400nm could be very difficult, which is not unique
>> to the imaging with the CSU system but universal to all optical and/or imaging
>> devices.
>> We have well confirmed the performance of the CSU confocal system including
>> all components such as the microscope, laser and fibers, all filters,
>> detector, etc.at 400-700 nm range.
>> However, we have no conclusive data on the use below 400nm range, in reality.
>>
>> Our engineers don't think there could be any simple and easy way to optimize
>> the CSU head for imaging at UV range, since every piece of optical parts both
>> inside and outside of the CSU head requires optimization for UV.
>> It may be possible to capture images below 400nm, but, we cannot guarantee
>> good-enough light efficiency, confocality or chromatic aberration correction,
>> and thus cannot endorse it.
>>
>> I am sorry for not being able to offer any quick solution but please kindly
>> understand there is no magic in good UV imaging.
>>
>>
>> Mizuho Shimizu
>>
>> Yokogawa Electric Corporation
>>
>>
>>
>>
>> -----Original Message-----
>> From: Confocal Microscopy List [mailto:[hidden email]] On
>> Behalf Of Renato Mortara
>> Sent: Monday, November 16, 2009 7:06 PM
>> To: [hidden email]
>> Subject: RES: Yokogawa lenses
>>
>> Mizuho
>> I hope you had a terrrific and relaxing weekend !
>>
>> Cheers
>>
>> Renato
>>
>> -----Mensagem original-----
>> De: Confocal Microscopy List [mailto:[hidden email]] Em nome
>> de [hidden email] Enviada em: sábado, 14 de novembro de 2009
>> 21:14
>> Para: [hidden email]
>> Assunto: Re: Yokogawa lenses
>>
>> Hi Alby, Guy, Renato and the list,
>>
>> Please allow us to "enjoy" a weekend holiday before getting back with
>> information to answer some of your questions regarding Yokogawa spinning disk.
>>
>> I wish everybody to have a nice holiday, too!
>>
>> Cheers,
>>
>> Mizuho
>>
>> ***********************************************************
>> Mizuho Shimizu
>> International Sales Team
>> BIO Group
>> B&A Center
>> Measurement Business Headquarters
>> Yokogawa Electric Corporation
>> 2-3 Hokuyodai, Kanazawa-shi, Ishikawa
>> 920-0177 JAPAN
>> Phone: +81-76-258-7028  Fax: +81-76-258-7029
>> <http://www.yokogawa.com/scanner/index.htm>
>> ***********************************************************
>> ________________________________________
>> 差出人: Confocal Microscopy List [[hidden email]] は
>> Renato A. Mortara [[hidden email]] の代理
>> 送信日時: 2009年11月15日 5:46
>> 宛先: [hidden email]
>> 件名: Re: Yokogawa lenses
>>
>> Alby and Guy,
>>
>> sure enough heating with UV is a concern.
>>
>> I did wish someone from Yokogawa read this list and gave a conclusive answer
>> whether any non UV-compatible element in the lightpath could be customized to
>> be UV friendly.
>>
>> Grazie and Ciao,
>>
>> Renato
>>
>>
>> Renato A. Mortara
>> Disciplina de Parasitologia
>> UNIFESP Escola Paulista de Medicina
>> R. Botucatu, 862 6o andar
>> 04023-062
>> S?o Paulo SP
>> Brasil
>>
>>
>> Quoting Alberto Diaspro <[hidden email]>:
>>
>>    
>>> It is possible that they also have to take into account the possible
>>> heating due to intense UV ciao Alby
>>>
>>>
>>>
>>>
>>> ISTITUTO ITALIANO
>>> DI TECNOLOGIA
>>>
>>> Prof. Alberto Diaspro
>>> Scientific Head
>>> Nanophysics
>>> Via Morego, 30 16163 Genova
>>> Tel: +39-010.71.781.503
>>> Fax +39-010-72.03.21
>>> Mobile +39-3666719968
>>> www.iit.it
>>>
>>>
>>> On Nov 14, 2009, at 12:55 PM, Guy Cox wrote:
>>>
>>>      
>>>> OK, from another commercial person (who does not want to be named)
>>>> I've now heard that in fact the Yokogawa microlenses are in fact
>>>> silica (as I had suspected).  But the snag is that there is another
>>>> optic in the system which does not pass UV.  This is a bit more
>>>> promising, since replacing the microlenses is not an option but
>>>> replacing a transfer lens shouldn't be difficult.  Someone from
>>>> Yokogawa must surely be reading this list.  One assumes, also, that
>>>> they would be  interested in extending the application range of their
>>>> system.  So, come on, Yokogawa, please give us the real story!
>>>>
>>>>                                                                Guy
>>>>
>>>>
>>>> Optical Imaging Techniques in Cell Biology
>>>> by Guy Cox    CRC Press / Taylor & Francis
>>>>    http://www.guycox.com/optical.htm
>>>> ______________________________________________
>>>> Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope
>>>> Unit, Madsen Building F09, University of Sydney, NSW 2006
>>>> ______________________________________________
>>>> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>>>> Mobile 0413 281 861
>>>> ______________________________________________
>>>>     http://www.guycox.net
>>>>
>>>>
>>>>
>>>>        
>>> ----------------------------------------
>>> Alberto Diaspro
>>> Head, Nanophysics Unit
>>> Senior Scientist
>>> The Italian Institute of Technology -IIT Via Morego, 30
>>> 16163 - Genova (Italy)
>>> phone: +39 010 71781503
>>> mobile: +393666719968
>>> fax:   +39 010 720321
>>> http://www.iit.it
>>> [hidden email]
>>>
>>> Professor of Applied Physics
>>> Department of Physics
>>> University of Genova
>>> Via Dodecaneso, 33
>>> 16146 Genova - Italy
>>> tel.  +39 010 353 6426
>>> fax. +39 010 314218
>>> http://www.lambs.it
>>> [hidden email]
>>> -------------------------------------------------------
>>>      
>> No virus found in this incoming message.
>> Checked by AVG - www.avg.com
>> Version: 9.0.707 / Virus Database: 270.14.61/2497 - Release Date: 11/17/09
>> 06:53:00
>>    
>
>  

--
Richard Berman
Spectral Applied Research
9078 Leslie St., Unit 11
Richmond Hill, Ontario
L4B 3L8

905-326-5040 ext. 444

www.spectral.ca
Renato A. Mortara Renato A. Mortara
Reply | Threaded
Open this post in threaded view
♦
♦
|

RES: RES: Yokogawa lenses

 

-----Mensagem original-----
De: Confocal Microscopy List [mailto:[hidden email]] Em nome de Richard Berman
Enviada em: terça-feira, 17 de novembro de 2009 20:32
Para: [hidden email]
Assunto: Re: RES: Yokogawa lenses

******** commercial interest *************

Hi Renato and Confocal List: As I stated in a previous post, we will be investigating the potential to operate the CSU at 375nm shortly. We expect the laser in about 2 weeks. As has been mentioned by another member, the potential to damage optics must always be considered. We do not believe that damage due to heating will be a factor since the heat is distributed over a fairly large area of the spinning disk. In fact if there is an issue it will be with the pinhole disk where the power densities are much higher. Damage due to UV is a greater concern but can be assessed. Many higher quality lenslet arrays are made from fused silica.  Transmission tests will enable us to determine if either of the disks are a polymer or both are fused silica. If both disks are silica they are probably not UV grade fused silica, but since we are not going to hard UV the standard grade will almost certainly show no degradation at 375nm. And by the way, single mode fibers that are rated to work at 375nm are made from the same fused silica as fibers for longer wavelengths. The power densities in the fibers are orders of magnitude higher than what will be hitting the disks and so degradation will be evident in the fiber long before the disks. This is something we are going to be assessing more carefully moving forward and should not be considered a guarantee of performance at this time.

And on a slightly different note, we have been very successful modifying the CSU to use lasers above 700nm, with emission wavelengths approaching 800nm. We have given the modifications a commercial name, BOREALIS. We are very excited about this development and our partner Quorum Technologies (also a commercial interest) has successfully installed two such systems with great results. Just to say that moving beyond the standard 400nm - 700nm range is possible.

Best Regards, Richard

Renato Mortara wrote:

> What is great in this list is that all the information flows along
> opinions and ideas. I agree with Guy as well as Paul. Unfortunately, I
> do not have enough $$ to play around with the systems,fibers,etc, just
> to try to implement an application... - my 2 cents in Brazilian Reais,
> ± $0.10
>
> Best
>
> Renato
>
> -----Mensagem original-----
> De: Confocal Microscopy List [mailto:[hidden email]]
> Em nome de Paul Maddox Enviada em: terça-feira, 17 de novembro de 2009
> 14:54
> Para: [hidden email]
> Assunto: Re: Yokogawa lenses
>
> Part of the "problem" here might be in way the question is being
> asked.  In my experience, the "end users" should take the initiative
> to experiment and try new applications of existing equipment.  Thereby
> we can empirically test these ideas on real world experimental
> applications.  However, to properly execute this, we need to know
> certain things such as if UV light (eg
> 300-400nm) will damage the system.  The question of if it will work or not is up to us to determine (ultimately by peer review).  The proper question to Yokogawa in this case is will we damage the system in doing so?
>
> Clearly, the greater the exchange of information the better for all involved.  But in practical terms, there is never a 100% free flow of info and I doubt there will be in the very near future.
>
> These are my two cents (Canadian, so ±1.8 US)...
>
> Paul
>
> Paul S. Maddox, PhD
> Assistant Professor
> Institute for Research in Immunology and Cancer Dept of Pathology and
> Cell Biol, U. de Montreal P.O. Box 6128, Station Centre-Ville Montréal
> QC  H3C 3J7 CANADA
>
> Courier:
> 2900, boulevard Édouard-Montpetit
> Pavillon Marcelle-Coutu, Quai 20
> Montreal QC  H3T 1J4
> CANADA
>
> [hidden email]
> Ph: 514-343-7894
> Fax: 514-343-6843
>
>
>
>
> On 17/11/09 11:17 AM, "Neeraj Gohad" <[hidden email]> wrote:
>
>  
>> I agree with Guy, there is much to be accomplished, the companies
>> should stopped treating scientist as just end users and we should
>> stop treating them as just vendors. Both parties will definitely
>> benefit if there is real dialogue.
>>
>> Best,
>>
>> Neeraj.
>>
>>
>> Neeraj V. Gohad, Ph.D.
>> Postdoctoral Fellow
>> Okeanos Research Group
>> Department of Biological Sciences
>> 132 Long Hall
>> Clemson University
>>
>> Please note my new email address: [hidden email]
>>
>>
>>
>>
>> -----Original Message-----
>> From: Confocal Microscopy List
>> [mailto:[hidden email]] On Behalf Of Guy Cox
>> Sent: Tuesday, November 17, 2009 6:18 AM
>> To: [hidden email]
>> Subject: Re: Yokogawa lenses
>>
>> Well, I'm glad your weekend was good (so was mine actually, with
>> kangaroos all around and lyre-birds playing in the garden).
>>
>> But your considered opinion after all that was to give us NO
>> information.  We are scientists, we are a community.  The
>> relationship between manufacturers and users is what drives advances in the microscopy field.  Please be a player
>> in this - it really is to your advantage.   What we need to know is simple -
>> something you can tell us in an instant.  What is the wavelength
>> transmission range of your micro-lenses?  For what wavelength are
>> they chromatically corrected?  What other components in your head
>> might affect the wavelength transmission?
>>
>> These are not great secrets of huge commercial value.  In fact,
>> someone will certainly make the relevant measurements and post them
>> if you don't.  But in the end collaboration is what drives progress,
>> so I hope you will change your mind and decide to become part of the community.
>>
>>
>>                  
>> Guy Cox
>>
>>
>>
>> Optical Imaging Techniques in Cell Biology
>> by Guy Cox    CRC Press / Taylor & Francis
>>     http://www.guycox.com/optical.htm 
>> ______________________________________________
>> Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope
>> Unit, Madsen Building F09, University of Sydney, NSW 2006
>> ______________________________________________
>> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>> Mobile 0413 281 861
>> ______________________________________________
>>      http://www.guycox.net
>> -----Original Message-----
>> From: Confocal Microscopy List
>> [mailto:[hidden email]] On Behalf Of
>> [hidden email]
>> Sent: Tuesday, 17 November 2009 12:20 PM
>> To: [hidden email]
>> Subject: Re: Yokogawa lenses
>>
>> Dear Renato, and the list.
>>
>> Yes, I enjoyed gorgeous autumn weather over the weekend!
>>
>>
>> While I am not in a position to officially represent Yokogawa, as
>> Renato has mentioned already, official comment from Yokogawa, if
>> requested, will be to ask to use the CSU spinning disk confocal
>> system within 400nm-700nm wavelength range for best performance with a good reason.
>>
>> As Lisa kindly advised, it is quite OK to use the CSUX1 system at 405 nm.
>> So, if your are talking about 405nm, it is within our specification range.
>>
>> However, if you are thinking about using  351/363 nm, as is well
>> understood, imaging at UV range below 400nm could be very difficult,
>> which is not unique to the imaging with the CSU system but universal
>> to all optical and/or imaging devices.
>> We have well confirmed the performance of the CSU confocal system
>> including all components such as the microscope, laser and fibers,
>> all filters, detector, etc.at 400-700 nm range.
>> However, we have no conclusive data on the use below 400nm range, in reality.
>>
>> Our engineers don't think there could be any simple and easy way to
>> optimize the CSU head for imaging at UV range, since every piece of
>> optical parts both inside and outside of the CSU head requires optimization for UV.
>> It may be possible to capture images below 400nm, but, we cannot
>> guarantee good-enough light efficiency, confocality or chromatic
>> aberration correction, and thus cannot endorse it.
>>
>> I am sorry for not being able to offer any quick solution but please
>> kindly understand there is no magic in good UV imaging.
>>
>>
>> Mizuho Shimizu
>>
>> Yokogawa Electric Corporation
>>
>>
>>
>>
>> -----Original Message-----
>> From: Confocal Microscopy List
>> [mailto:[hidden email]] On Behalf Of Renato Mortara
>> Sent: Monday, November 16, 2009 7:06 PM
>> To: [hidden email]
>> Subject: RES: Yokogawa lenses
>>
>> Mizuho
>> I hope you had a terrrific and relaxing weekend !
>>
>> Cheers
>>
>> Renato
>>
>> -----Mensagem original-----
>> De: Confocal Microscopy List
>> [mailto:[hidden email]] Em nome de
>> [hidden email] Enviada em: sábado, 14 de novembro de
>> 2009
>> 21:14
>> Para: [hidden email]
>> Assunto: Re: Yokogawa lenses
>>
>> Hi Alby, Guy, Renato and the list,
>>
>> Please allow us to "enjoy" a weekend holiday before getting back with
>> information to answer some of your questions regarding Yokogawa spinning disk.
>>
>> I wish everybody to have a nice holiday, too!
>>
>> Cheers,
>>
>> Mizuho
>>
>> ***********************************************************
>> Mizuho Shimizu
>> International Sales Team
>> BIO Group
>> B&A Center
>> Measurement Business Headquarters
>> Yokogawa Electric Corporation
>> 2-3 Hokuyodai, Kanazawa-shi, Ishikawa
>> 920-0177 JAPAN
>> Phone: +81-76-258-7028  Fax: +81-76-258-7029
>> <http://www.yokogawa.com/scanner/index.htm>
>> ***********************************************************
>> ________________________________________
>> 差出人: Confocal Microscopy List [[hidden email]] は
>> Renato A. Mortara [[hidden email]] の代理
>> 送信日時: 2009年11月15日 5:46
>> 宛先: [hidden email]
>> 件名: Re: Yokogawa lenses
>>
>> Alby and Guy,
>>
>> sure enough heating with UV is a concern.
>>
>> I did wish someone from Yokogawa read this list and gave a conclusive
>> answer whether any non UV-compatible element in the lightpath could
>> be customized to be UV friendly.
>>
>> Grazie and Ciao,
>>
>> Renato
>>
>>
>> Renato A. Mortara
>> Disciplina de Parasitologia
>> UNIFESP Escola Paulista de Medicina
>> R. Botucatu, 862 6o andar
>> 04023-062
>> S?o Paulo SP
>> Brasil
>>
>>
>> Quoting Alberto Diaspro <[hidden email]>:
>>
>>    
>>> It is possible that they also have to take into account the possible
>>> heating due to intense UV ciao Alby
>>>
>>>
>>>
>>>
>>> ISTITUTO ITALIANO
>>> DI TECNOLOGIA
>>>
>>> Prof. Alberto Diaspro
>>> Scientific Head
>>> Nanophysics
>>> Via Morego, 30 16163 Genova
>>> Tel: +39-010.71.781.503
>>> Fax +39-010-72.03.21
>>> Mobile +39-3666719968
>>> www.iit.it
>>>
>>>
>>> On Nov 14, 2009, at 12:55 PM, Guy Cox wrote:
>>>
>>>      
>>>> OK, from another commercial person (who does not want to be named)
>>>> I've now heard that in fact the Yokogawa microlenses are in fact
>>>> silica (as I had suspected).  But the snag is that there is another
>>>> optic in the system which does not pass UV.  This is a bit more
>>>> promising, since replacing the microlenses is not an option but
>>>> replacing a transfer lens shouldn't be difficult.  Someone from
>>>> Yokogawa must surely be reading this list.  One assumes, also, that
>>>> they would be  interested in extending the application range of
>>>> their system.  So, come on, Yokogawa, please give us the real story!
>>>>
>>>>                                                                Guy
>>>>
>>>>
>>>> Optical Imaging Techniques in Cell Biology
>>>> by Guy Cox    CRC Press / Taylor & Francis
>>>>    http://www.guycox.com/optical.htm 
>>>> ______________________________________________
>>>> Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope
>>>> Unit, Madsen Building F09, University of Sydney, NSW 2006
>>>> ______________________________________________
>>>> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>>>> Mobile 0413 281 861
>>>> ______________________________________________
>>>>     http://www.guycox.net
>>>>
>>>>
>>>>
>>>>        
>>> ----------------------------------------
>>> Alberto Diaspro
>>> Head, Nanophysics Unit
>>> Senior Scientist
>>> The Italian Institute of Technology -IIT Via Morego, 30
>>> 16163 - Genova (Italy)
>>> phone: +39 010 71781503
>>> mobile: +393666719968
>>> fax:   +39 010 720321
>>> http://www.iit.it
>>> [hidden email]
>>>
>>> Professor of Applied Physics
>>> Department of Physics
>>> University of Genova
>>> Via Dodecaneso, 33
>>> 16146 Genova - Italy
>>> tel.  +39 010 353 6426
>>> fax. +39 010 314218
>>> http://www.lambs.it
>>> [hidden email]
>>> -------------------------------------------------------
>>>      
>> No virus found in this incoming message.
>> Checked by AVG - www.avg.com
>> Version: 9.0.707 / Virus Database: 270.14.61/2497 - Release Date:
>> 11/17/09 06:53:00
>>    
>
>  

--
Richard Berman
Spectral Applied Research
9078 Leslie St., Unit 11
Richmond Hill, Ontario
L4B 3L8

905-326-5040 ext. 444

www.spectral.ca
Renato A. Mortara Renato A. Mortara
Reply | Threaded
Open this post in threaded view
♦
♦
|

RES: RES: Yokogawa lenses

In reply to this post by Richard Berman
Hi Richard and list members,

It is great that you will be able to bench test UV lines on the Yokogawa head and fibers.

Keeping my fingers crossed for a positive outcome !

Best

Renato

 

-----Mensagem original-----
De: Confocal Microscopy List [mailto:[hidden email]] Em nome de Richard Berman
Enviada em: terça-feira, 17 de novembro de 2009 20:32
Para: [hidden email]
Assunto: Re: RES: Yokogawa lenses

******** commercial interest *************

Hi Renato and Confocal List: As I stated in a previous post, we will be investigating the potential to operate the CSU at 375nm shortly. We expect the laser in about 2 weeks. As has been mentioned by another member, the potential to damage optics must always be considered. We do not believe that damage due to heating will be a factor since the heat is distributed over a fairly large area of the spinning disk. In fact if there is an issue it will be with the pinhole disk where the power densities are much higher. Damage due to UV is a greater concern but can be assessed. Many higher quality lenslet arrays are made from fused silica.  Transmission tests will enable us to determine if either of the disks are a polymer or both are fused silica. If both disks are silica they are probably not UV grade fused silica, but since we are not going to hard UV the standard grade will almost certainly show no degradation at 375nm. And by the way, single mode fibers that are rated to work at 375nm are made from the same fused silica as fibers for longer wavelengths. The power densities in the fibers are orders of magnitude higher than what will be hitting the disks and so degradation will be evident in the fiber long before the disks. This is something we are going to be assessing more carefully moving forward and should not be considered a guarantee of performance at this time.

And on a slightly different note, we have been very successful modifying the CSU to use lasers above 700nm, with emission wavelengths approaching 800nm. We have given the modifications a commercial name, BOREALIS. We are very excited about this development and our partner Quorum Technologies (also a commercial interest) has successfully installed two such systems with great results. Just to say that moving beyond the standard 400nm - 700nm range is possible.

Best Regards, Richard

Renato Mortara wrote:

> What is great in this list is that all the information flows along
> opinions and ideas. I agree with Guy as well as Paul. Unfortunately, I
> do not have enough $$ to play around with the systems,fibers,etc, just
> to try to implement an application... - my 2 cents in Brazilian Reais,
> ± $0.10
>
> Best
>
> Renato
>
> -----Mensagem original-----
> De: Confocal Microscopy List [mailto:[hidden email]]
> Em nome de Paul Maddox Enviada em: terça-feira, 17 de novembro de 2009
> 14:54
> Para: [hidden email]
> Assunto: Re: Yokogawa lenses
>
> Part of the "problem" here might be in way the question is being
> asked.  In my experience, the "end users" should take the initiative
> to experiment and try new applications of existing equipment.  Thereby
> we can empirically test these ideas on real world experimental
> applications.  However, to properly execute this, we need to know
> certain things such as if UV light (eg
> 300-400nm) will damage the system.  The question of if it will work or not is up to us to determine (ultimately by peer review).  The proper question to Yokogawa in this case is will we damage the system in doing so?
>
> Clearly, the greater the exchange of information the better for all involved.  But in practical terms, there is never a 100% free flow of info and I doubt there will be in the very near future.
>
> These are my two cents (Canadian, so ±1.8 US)...
>
> Paul
>
> Paul S. Maddox, PhD
> Assistant Professor
> Institute for Research in Immunology and Cancer Dept of Pathology and
> Cell Biol, U. de Montreal P.O. Box 6128, Station Centre-Ville Montréal
> QC  H3C 3J7 CANADA
>
> Courier:
> 2900, boulevard Édouard-Montpetit
> Pavillon Marcelle-Coutu, Quai 20
> Montreal QC  H3T 1J4
> CANADA
>
> [hidden email]
> Ph: 514-343-7894
> Fax: 514-343-6843
>
>
>
>
> On 17/11/09 11:17 AM, "Neeraj Gohad" <[hidden email]> wrote:
>
>  
>> I agree with Guy, there is much to be accomplished, the companies
>> should stopped treating scientist as just end users and we should
>> stop treating them as just vendors. Both parties will definitely
>> benefit if there is real dialogue.
>>
>> Best,
>>
>> Neeraj.
>>
>>
>> Neeraj V. Gohad, Ph.D.
>> Postdoctoral Fellow
>> Okeanos Research Group
>> Department of Biological Sciences
>> 132 Long Hall
>> Clemson University
>>
>> Please note my new email address: [hidden email]
>>
>>
>>
>>
>> -----Original Message-----
>> From: Confocal Microscopy List
>> [mailto:[hidden email]] On Behalf Of Guy Cox
>> Sent: Tuesday, November 17, 2009 6:18 AM
>> To: [hidden email]
>> Subject: Re: Yokogawa lenses
>>
>> Well, I'm glad your weekend was good (so was mine actually, with
>> kangaroos all around and lyre-birds playing in the garden).
>>
>> But your considered opinion after all that was to give us NO
>> information.  We are scientists, we are a community.  The
>> relationship between manufacturers and users is what drives advances in the microscopy field.  Please be a player
>> in this - it really is to your advantage.   What we need to know is simple -
>> something you can tell us in an instant.  What is the wavelength
>> transmission range of your micro-lenses?  For what wavelength are
>> they chromatically corrected?  What other components in your head
>> might affect the wavelength transmission?
>>
>> These are not great secrets of huge commercial value.  In fact,
>> someone will certainly make the relevant measurements and post them
>> if you don't.  But in the end collaboration is what drives progress,
>> so I hope you will change your mind and decide to become part of the community.
>>
>>
>>                  
>> Guy Cox
>>
>>
>>
>> Optical Imaging Techniques in Cell Biology
>> by Guy Cox    CRC Press / Taylor & Francis
>>     http://www.guycox.com/optical.htm 
>> ______________________________________________
>> Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope
>> Unit, Madsen Building F09, University of Sydney, NSW 2006
>> ______________________________________________
>> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>> Mobile 0413 281 861
>> ______________________________________________
>>      http://www.guycox.net
>> -----Original Message-----
>> From: Confocal Microscopy List
>> [mailto:[hidden email]] On Behalf Of
>> [hidden email]
>> Sent: Tuesday, 17 November 2009 12:20 PM
>> To: [hidden email]
>> Subject: Re: Yokogawa lenses
>>
>> Dear Renato, and the list.
>>
>> Yes, I enjoyed gorgeous autumn weather over the weekend!
>>
>>
>> While I am not in a position to officially represent Yokogawa, as
>> Renato has mentioned already, official comment from Yokogawa, if
>> requested, will be to ask to use the CSU spinning disk confocal
>> system within 400nm-700nm wavelength range for best performance with a good reason.
>>
>> As Lisa kindly advised, it is quite OK to use the CSUX1 system at 405 nm.
>> So, if your are talking about 405nm, it is within our specification range.
>>
>> However, if you are thinking about using  351/363 nm, as is well
>> understood, imaging at UV range below 400nm could be very difficult,
>> which is not unique to the imaging with the CSU system but universal
>> to all optical and/or imaging devices.
>> We have well confirmed the performance of the CSU confocal system
>> including all components such as the microscope, laser and fibers,
>> all filters, detector, etc.at 400-700 nm range.
>> However, we have no conclusive data on the use below 400nm range, in reality.
>>
>> Our engineers don't think there could be any simple and easy way to
>> optimize the CSU head for imaging at UV range, since every piece of
>> optical parts both inside and outside of the CSU head requires optimization for UV.
>> It may be possible to capture images below 400nm, but, we cannot
>> guarantee good-enough light efficiency, confocality or chromatic
>> aberration correction, and thus cannot endorse it.
>>
>> I am sorry for not being able to offer any quick solution but please
>> kindly understand there is no magic in good UV imaging.
>>
>>
>> Mizuho Shimizu
>>
>> Yokogawa Electric Corporation
>>
>>
>>
>>
>> -----Original Message-----
>> From: Confocal Microscopy List
>> [mailto:[hidden email]] On Behalf Of Renato Mortara
>> Sent: Monday, November 16, 2009 7:06 PM
>> To: [hidden email]
>> Subject: RES: Yokogawa lenses
>>
>> Mizuho
>> I hope you had a terrrific and relaxing weekend !
>>
>> Cheers
>>
>> Renato
>>
>> -----Mensagem original-----
>> De: Confocal Microscopy List
>> [mailto:[hidden email]] Em nome de
>> [hidden email] Enviada em: sábado, 14 de novembro de
>> 2009
>> 21:14
>> Para: [hidden email]
>> Assunto: Re: Yokogawa lenses
>>
>> Hi Alby, Guy, Renato and the list,
>>
>> Please allow us to "enjoy" a weekend holiday before getting back with
>> information to answer some of your questions regarding Yokogawa spinning disk.
>>
>> I wish everybody to have a nice holiday, too!
>>
>> Cheers,
>>
>> Mizuho
>>
>> ***********************************************************
>> Mizuho Shimizu
>> International Sales Team
>> BIO Group
>> B&A Center
>> Measurement Business Headquarters
>> Yokogawa Electric Corporation
>> 2-3 Hokuyodai, Kanazawa-shi, Ishikawa
>> 920-0177 JAPAN
>> Phone: +81-76-258-7028  Fax: +81-76-258-7029
>> <http://www.yokogawa.com/scanner/index.htm>
>> ***********************************************************
>> ________________________________________
>> 差出人: Confocal Microscopy List [[hidden email]] は
>> Renato A. Mortara [[hidden email]] の代理
>> 送信日時: 2009年11月15日 5:46
>> 宛先: [hidden email]
>> 件名: Re: Yokogawa lenses
>>
>> Alby and Guy,
>>
>> sure enough heating with UV is a concern.
>>
>> I did wish someone from Yokogawa read this list and gave a conclusive
>> answer whether any non UV-compatible element in the lightpath could
>> be customized to be UV friendly.
>>
>> Grazie and Ciao,
>>
>> Renato
>>
>>
>> Renato A. Mortara
>> Disciplina de Parasitologia
>> UNIFESP Escola Paulista de Medicina
>> R. Botucatu, 862 6o andar
>> 04023-062
>> S?o Paulo SP
>> Brasil
>>
>>
>> Quoting Alberto Diaspro <[hidden email]>:
>>
>>    
>>> It is possible that they also have to take into account the possible
>>> heating due to intense UV ciao Alby
>>>
>>>
>>>
>>>
>>> ISTITUTO ITALIANO
>>> DI TECNOLOGIA
>>>
>>> Prof. Alberto Diaspro
>>> Scientific Head
>>> Nanophysics
>>> Via Morego, 30 16163 Genova
>>> Tel: +39-010.71.781.503
>>> Fax +39-010-72.03.21
>>> Mobile +39-3666719968
>>> www.iit.it
>>>
>>>
>>> On Nov 14, 2009, at 12:55 PM, Guy Cox wrote:
>>>
>>>      
>>>> OK, from another commercial person (who does not want to be named)
>>>> I've now heard that in fact the Yokogawa microlenses are in fact
>>>> silica (as I had suspected).  But the snag is that there is another
>>>> optic in the system which does not pass UV.  This is a bit more
>>>> promising, since replacing the microlenses is not an option but
>>>> replacing a transfer lens shouldn't be difficult.  Someone from
>>>> Yokogawa must surely be reading this list.  One assumes, also, that
>>>> they would be  interested in extending the application range of
>>>> their system.  So, come on, Yokogawa, please give us the real story!
>>>>
>>>>                                                                Guy
>>>>
>>>>
>>>> Optical Imaging Techniques in Cell Biology
>>>> by Guy Cox    CRC Press / Taylor & Francis
>>>>    http://www.guycox.com/optical.htm 
>>>> ______________________________________________
>>>> Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope
>>>> Unit, Madsen Building F09, University of Sydney, NSW 2006
>>>> ______________________________________________
>>>> Phone +61 2 9351 3176     Fax +61 2 9351 7682
>>>> Mobile 0413 281 861
>>>> ______________________________________________
>>>>     http://www.guycox.net
>>>>
>>>>
>>>>
>>>>        
>>> ----------------------------------------
>>> Alberto Diaspro
>>> Head, Nanophysics Unit
>>> Senior Scientist
>>> The Italian Institute of Technology -IIT Via Morego, 30
>>> 16163 - Genova (Italy)
>>> phone: +39 010 71781503
>>> mobile: +393666719968
>>> fax:   +39 010 720321
>>> http://www.iit.it
>>> [hidden email]
>>>
>>> Professor of Applied Physics
>>> Department of Physics
>>> University of Genova
>>> Via Dodecaneso, 33
>>> 16146 Genova - Italy
>>> tel.  +39 010 353 6426
>>> fax. +39 010 314218
>>> http://www.lambs.it
>>> [hidden email]
>>> -------------------------------------------------------
>>>      
>> No virus found in this incoming message.
>> Checked by AVG - www.avg.com
>> Version: 9.0.707 / Virus Database: 270.14.61/2497 - Release Date:
>> 11/17/09 06:53:00
>>    
>
>  

--
Richard Berman
Spectral Applied Research
9078 Leslie St., Unit 11
Richmond Hill, Ontario
L4B 3L8

905-326-5040 ext. 444

www.spectral.ca