To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** We have a 2008 model Axioimager, coupled to an LSM5 exciter, used mainly for polarization studies on liquid crystals. Recently, the Zen software failed to launch with "fail to connect to RTC..." message. Resetting did not work. On the monitor, however, 'Real time connection connected..' message appears. It is suspected that the IDE, transcend 128 MB flash module, used to boot may have been corrupted. If a new IDE flash module is purchased, the program data will have to be written into it. Is there a way of getting the data? Since the model is out of service, local Zeiss rep has expressed inability to extend technical support. It seems such a pity that for the sake of the IDE, the instrument should remain defunct. We would be indebted for any suggestions to bring the confocal onroad. |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** We can help. Www.Arcroyal.com ________________________________ From: Confocal Microscopy List <[hidden email]> on behalf of Prof. Dr. K. S. Krishnamurthy <[hidden email]> Sent: Saturday, June 27, 2020 11:52:36 PM To: [hidden email] <[hidden email]> Subject: Failed RTC connection ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** We have a 2008 model Axioimager, coupled to an LSM5 exciter, used mainly for polarization studies on liquid crystals. Recently, the Zen software failed to launch with "fail to connect to RTC..." message. Resetting did not work. On the monitor, however, 'Real time connection connected..' message appears. It is suspected that the IDE, transcend 128 MB flash module, used to boot may have been corrupted. If a new IDE flash module is purchased, the program data will have to be written into it. Is there a way of getting the data? Since the model is out of service, local Zeiss rep has expressed inability to extend technical support. It seems such a pity that for the sake of the IDE, the instrument should remain defunct. We would be indebted for any suggestions to bring the confocal onroad. |
|
|
In reply to this post by Prof. Dr. K. S. Krishnamurthy
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Krishnamurthy, I don't know if all the RTC PCs have the same operative system (our machine is a LSM7DUO). If so, I can make a copy of our flash drive and send you the compressed files, so you can rebuild your system again. On the other hand, thank you for remembering this black box computer needs also an HD backup. Best regards. Juan Luis El 28/06/2020 a las 6:52, Prof. Dr. K. S. Krishnamurthy escribió: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > We have a 2008 model Axioimager, coupled to an LSM5 exciter, used mainly for polarization studies on liquid crystals. Recently, the Zen software failed to launch with "fail to connect to RTC..." message. Resetting did not work. On the monitor, however, 'Real time connection connected..' message appears. It is suspected that the IDE, transcend 128 MB flash module, used to boot may have been corrupted. If a new IDE flash module is purchased, the program data will have to be written into it. Is there a way of getting the data? Since the model is out of service, local Zeiss rep has expressed inability to extend technical support. It seems such a pity that for the sake of the IDE, the instrument should remain defunct. We would be indebted for any suggestions to bring the confocal onroad. -- /Juan Luis Ribas/ Servicio de Microscopía Centro de Investigación, Tecnología e Innovación Universidad de Sevilla Av. Reina Mercedes 4b 41012 Sevilla Tfnos: 954559733 // 628220681 // 12068 ¿Cómo llegar? <https://goo.gl/maps/hxw4JLxKpTA2> Equipos <http://citius.us.es/web/servicio.php?s=mic#equipos> / Altas <http://citius.us.es/web/servicio.php?s=mic#documentos> / Tarifas <http://citius.us.es/web/servicio.php?s=mic#documentos> / Reservas <http://alojaservicios.us.es/microscopia/php/OCFspanish/select_resource.php?v=1> |
|
|
In reply to this post by Prof. Dr. K. S. Krishnamurthy
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Good morning, I don't know if this will help but writing to you just in case the problem is less serious than you think. We have a Zeiss LSM 710. After we moved to Windows 7 some years ago, we had to change the order in which we turned things on. If we turn on the RTC before the PC, we get the same error. Using the Reset button doesn't work, we just have to turn off the RTC, exit ZEN, restart the RTC and wait for the start up sounds Then restart ZEN and all is OK. We changed our practice to turning on the RTC after the PC is on and after we have logged on. This solved the problem. Also our system has a back up copy of the system on a USB that is stored inside the PC. I think this is standard for Zeiss so you should also have this. Kind regards, Jacqui ________________________________________ From: Confocal Microscopy List [[hidden email]] on behalf of Prof. Dr. K. S. Krishnamurthy [[hidden email]] Sent: 28 June 2020 16:52 To: [hidden email] Subject: Failed RTC connection ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** We have a 2008 model Axioimager, coupled to an LSM5 exciter, used mainly for polarization studies on liquid crystals. Recently, the Zen software failed to launch with "fail to connect to RTC..." message. Resetting did not work. On the monitor, however, 'Real time connection connected..' message appears. It is suspected that the IDE, transcend 128 MB flash module, used to boot may have been corrupted. If a new IDE flash module is purchased, the program data will have to be written into it. Is there a way of getting the data? Since the model is out of service, local Zeiss rep has expressed inability to extend technical support. It seems such a pity that for the sake of the IDE, the instrument should remain defunct. We would be indebted for any suggestions to bring the confocal onroad. |
|
G. Esteban Fernandez |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** In case it helps, I was just having RTC connection issues on an LSM 710 and it turns out the RTC was not booting up correctly for lack of a keyboard (we think BIOS was reset). Connecting a PS/2 keyboard to the back of the RTC fixed the issue while Zeiss resets the BIOS and replaces the motherboard battery. -Esteban On Sun, Jun 28, 2020 at 12:15 PM Jacqueline Ross <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Good morning, > > I don't know if this will help but writing to you just in case the problem > is less serious than you think. > > We have a Zeiss LSM 710. After we moved to Windows 7 some years ago, we > had to change the order in which we turned things on. If we turn on the RTC > before the PC, we get the same error. Using the Reset button doesn't work, > we just have to turn off the RTC, exit ZEN, restart the RTC and wait for > the start up sounds Then restart ZEN and all is OK. > > We changed our practice to turning on the RTC after the PC is on and after > we have logged on. This solved the problem. > > Also our system has a back up copy of the system on a USB that is stored > inside the PC. I think this is standard for Zeiss so you should also have > this. > > Kind regards, > > Jacqui > ________________________________________ > From: Confocal Microscopy List [[hidden email]] on > behalf of Prof. Dr. K. S. Krishnamurthy [[hidden email]] > Sent: 28 June 2020 16:52 > To: [hidden email] > Subject: Failed RTC connection > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > We have a 2008 model Axioimager, coupled to an LSM5 exciter, used mainly > for polarization studies on liquid crystals. Recently, the Zen software > failed to launch with "fail to connect to RTC..." message. Resetting did > not work. On the monitor, however, 'Real time connection connected..' > message appears. It is suspected that the IDE, transcend 128 MB flash > module, used to boot may have been corrupted. If a new IDE flash module is > purchased, the program data will have to be written into it. Is there a way > of getting the data? Since the model is out of service, local Zeiss rep has > expressed inability to extend technical support. It seems such a pity that > for the sake of the IDE, the instrument should remain defunct. We would be > indebted for any suggestions to bring the confocal onroad. > |
|
Prof. Dr. K. S. Krishnamurthy |
|
|
In reply to this post by Prof. Dr. K. S. Krishnamurthy
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Good morning. Thank you for the input. Our computer engineer tried using the keyboard, but did not succeed in booting. |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi , Maybe it had been mentionned before: 1- change the battery of the motiver Board, reprogram the Bios , resynchronise the Clock, date between the RTC and the PC 2- check the connection to the motherboard Good luck Louis Téléchargez Outlook pour iOS<https://aka.ms/o0ukef> ________________________________ De : Confocal Microscopy List <[hidden email]> de la part de Prof. Dr. K. S. Krishnamurthy <[hidden email]> Envoyé : Monday, June 29, 2020 12:21:50 AM À : [hidden email] <[hidden email]> Objet : Re: Failed RTC connection Avertissement automatisé : Ce courriel provient de l'extérieur de votre organisation. Ne cliquez pas sur les liens et les pièces jointes si vous ne reconnaissez pas l'expéditeur. ***** To join, leave or search the confocal microscopy listserv, go to: https://can01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01%7Clouis.villeneuve%40ICM-MHI.ORG%7Cb0f0d51dd4e1435f3a3408d81be61332%7C06e1fe285f8b4075bf6cae24be1a7992%7C0%7C0%7C637290022863641065&sdata=XfyPooyK3yFThct1tD8wsMUzly2k7ewzoxGQZjDghW4%3D&reserved=0 Post images on https://can01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&data=02%7C01%7Clouis.villeneuve%40ICM-MHI.ORG%7Cb0f0d51dd4e1435f3a3408d81be61332%7C06e1fe285f8b4075bf6cae24be1a7992%7C0%7C0%7C637290022863641065&sdata=GAZ6Huf0H1cy9rE4u5rNu2nbj%2FzCMJDGh3Svd80WyWA%3D&reserved=0 and include the link in your posting. ***** Good morning. Thank you for the input. Our computer engineer tried using the keyboard, but did not succeed in booting. |
|
G. Esteban Fernandez |
|
|
In reply to this post by Prof. Dr. K. S. Krishnamurthy
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** You can also connect a monitor to the RTC to see what it’s doing and if it gives a clue to what to try. -Esteban On Sun, Jun 28, 2020 at 9:37 PM Prof. Dr. K. S. Krishnamurthy < [hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Good morning. Thank you for the input. Our computer engineer tried using > the keyboard, but did not succeed in booting. > |
|
Prof. Dr. K. S. Krishnamurthy |
|
|
In reply to this post by Louis Villeneuve
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Dr Louis, Thank you for the suggestions. We tried booting the RTC using a keyboard and monitor. We get an error message, which I have loaded to Google drive: https://drive.google.com/file/d/1OaWJKOEufbFJTz4_wbHNDWrmcViuGtZw/view?usp=sharing Any suggestions? Regards Krishnamurthy > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi , Maybe it had been mentionned before: > > 1- change the battery of the motiver Board, reprogram the Bios , > resynchronise the Clock, date between the RTC and the PC > > 2- check the connection to the motherboard > > Good luck > > Louis > > > Téléchargez Outlook pour iOS<https://aka.ms/o0ukef> > ________________________________ > De : Confocal Microscopy List <[hidden email]> de la > part de Prof. Dr. K. S. Krishnamurthy <[hidden email]> > Envoyé : Monday, June 29, 2020 12:21:50 AM > À : [hidden email] <[hidden email]> > Objet : Re: Failed RTC connection > > > Avertissement automatisé : Ce courriel provient de l'extérieur de votre > organisation. Ne cliquez pas sur les liens et les pièces jointes si vous > ne reconnaissez pas l'expéditeur. > > > > ***** > To join, leave or search the confocal microscopy listserv, go to: > https://can01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01%7Clouis.villeneuve%40ICM-MHI.ORG%7Cb0f0d51dd4e1435f3a3408d81be61332%7C06e1fe285f8b4075bf6cae24be1a7992%7C0%7C0%7C637290022863641065&sdata=XfyPooyK3yFThct1tD8wsMUzly2k7ewzoxGQZjDghW4%3D&reserved=0 > Post images on > https://can01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&data=02%7C01%7Clouis.villeneuve%40ICM-MHI.ORG%7Cb0f0d51dd4e1435f3a3408d81be61332%7C06e1fe285f8b4075bf6cae24be1a7992%7C0%7C0%7C637290022863641065&sdata=GAZ6Huf0H1cy9rE4u5rNu2nbj%2FzCMJDGh3Svd80WyWA%3D&reserved=0 > and include the link in your posting. > ***** > > Good morning. Thank you for the input. Our computer engineer tried using > the keyboard, but did not succeed in booting. > -- Centre for Nano and Soft Matter Sciences PO Box 1329 Jalahalli bangalore 560 013 Ph: +91 80 23084236 (O), Cell:9845284587 |
|
Prof. Dr. K. S. Krishnamurthy |
|
|
In reply to this post by G. Esteban Fernandez
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Dr Estaban, Good morning. We tried using a monitor. We get the error message shown in the image uploaded to my google drive: https://drive.google.com/file/d/1OaWJKOEufbFJTz4_wbHNDWrmcViuGtZw/view?usp=sharing Any suggestions? Regards Krishnamurthy > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > You can also connect a monitor to the RTC to see what itâs doing and if > it > gives a clue to what to try. > > -Esteban > > > On Sun, Jun 28, 2020 at 9:37 PM Prof. Dr. K. S. Krishnamurthy < > [hidden email]> wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your >> posting. >> ***** >> >> Good morning. Thank you for the input. Our computer engineer tried using >> the keyboard, but did not succeed in booting. >> > -- Centre for Nano and Soft Matter Sciences PO Box 1329 Jalahalli bangalore 560 013 Ph: +91 80 23084236 (O), Cell:9845284587 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** That looks like some sort of drive failure. BLK_DEV is basically the description of the hardware and how to use it, so if you can't initiate it then the hardware might not be responding to the bootloader/BIOS. Craig On Mon, Jun 29, 2020 at 10:32 PM Prof. K. S. Krishnamurthy < [hidden email]> wrote: > ***** > Dear Dr Estaban, > Good morning. > We tried using a monitor. We get the error message shown in the image > uploaded to my google drive: > > https://drive.google.com/file/d/1OaWJKOEufbFJTz4_wbHNDWrmcViuGtZw/view?usp=sharing > Any suggestions? > Regards > Krishnamurthy > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > You can also connect a monitor to the RTC to see what it’s doing and if > > it > > gives a clue to what to try. > > > > -Esteban > > > > > > On Sun, Jun 28, 2020 at 9:37 PM Prof. Dr. K. S. Krishnamurthy < > > [hidden email]> wrote: > > > >> ***** > >> To join, leave or search the confocal microscopy listserv, go to: > >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >> Post images on http://www.imgur.com and include the link in your > >> posting. > >> ***** > >> > >> Good morning. Thank you for the input. Our computer engineer tried using > >> the keyboard, but did not succeed in booting. > >> > > > > > -- > Centre for Nano and Soft Matter Sciences > PO Box 1329 > Jalahalli > bangalore 560 013 > Ph: +91 80 23084236 (O), Cell:9845284587 > |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** There won't be many VxWorks experts around here... But yes, this looks like a corrupted drive. It could also be BIOS configuration, I would definitely try to toggle between IDE and AHCI settings in BIOS (provided the drive is connected through SATA). Another thought: Zeiss might have an easy way to recover the RT computer, maybe TFTP/BOOTP, but this will be hard to figure out without Zeiss' help. Good luck! zdenek On Tue, Jun 30, 2020 at 12:42 AM Craig Brideau <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > That looks like some sort of drive failure. BLK_DEV is basically the > description of the hardware and how to use it, so if you can't initiate it > then the hardware might not be responding to the bootloader/BIOS. > > Craig > > On Mon, Jun 29, 2020 at 10:32 PM Prof. K. S. Krishnamurthy < > [hidden email]> wrote: > > > ***** > > Dear Dr Estaban, > > Good morning. > > We tried using a monitor. We get the error message shown in the image > > uploaded to my google drive: > > > > > https://drive.google.com/file/d/1OaWJKOEufbFJTz4_wbHNDWrmcViuGtZw/view?usp=sharing > > Any suggestions? > > Regards > > Krishnamurthy > > > > > ***** > > > To join, leave or search the confocal microscopy listserv, go to: > > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > > Post images on http://www.imgur.com and include the link in your > > posting. > > > ***** > > > > > > You can also connect a monitor to the RTC to see what it’s doing and if > > > it > > > gives a clue to what to try. > > > > > > -Esteban > > > > > > > > > On Sun, Jun 28, 2020 at 9:37 PM Prof. Dr. K. S. Krishnamurthy < > > > [hidden email]> wrote: > > > > > >> ***** > > >> To join, leave or search the confocal microscopy listserv, go to: > > >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > >> Post images on http://www.imgur.com and include the link in your > > >> posting. > > >> ***** > > >> > > >> Good morning. Thank you for the input. Our computer engineer tried > using > > >> the keyboard, but did not succeed in booting. > > >> > > > > > > > > > -- > > Centre for Nano and Soft Matter Sciences > > PO Box 1329 > > Jalahalli > > bangalore 560 013 > > Ph: +91 80 23084236 (O), Cell:9845284587 > > > -- -- Zdenek Svindrych, Ph.D. Research Scientist - Microscopy Imaging Specialist Department of Biochemistry and Cell Biology Geisel School of Medicine at Dartmouth |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi, we had this problem on our 510 some years back when I was working in Norway. I posted the solution on the confocal list then. Here is the link: http://confocal-microscopy-list.588098.n2.nabble.com/How-to-repair-a-broken-RTC-on-a-Zeiss-LSM-510-td7583065.html Jens Eriksson Uppsala university On Tue, 30 Jun 2020, 16:47 Zdenek Svindrych, <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > There won't be many VxWorks experts around here... But yes, this looks like > a corrupted drive. > It could also be BIOS configuration, I would definitely try to toggle > between IDE and AHCI settings in BIOS (provided the drive is connected > through SATA). > Another thought: Zeiss might have an easy way to recover the RT computer, > maybe TFTP/BOOTP, but this will be hard to figure out without Zeiss' help. > Good luck! > zdenek > > On Tue, Jun 30, 2020 at 12:42 AM Craig Brideau <[hidden email]> > wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > That looks like some sort of drive failure. BLK_DEV is basically the > > description of the hardware and how to use it, so if you can't initiate > it > > then the hardware might not be responding to the bootloader/BIOS. > > > > Craig > > > > On Mon, Jun 29, 2020 at 10:32 PM Prof. K. S. Krishnamurthy < > > [hidden email]> wrote: > > > > > ***** > > > Dear Dr Estaban, > > > Good morning. > > > We tried using a monitor. We get the error message shown in the image > > > uploaded to my google drive: > > > > > > > > > https://drive.google.com/file/d/1OaWJKOEufbFJTz4_wbHNDWrmcViuGtZw/view?usp=sharing > > > Any suggestions? > > > Regards > > > Krishnamurthy > > > > > > > ***** > > > > To join, leave or search the confocal microscopy listserv, go to: > > > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > > > Post images on http://www.imgur.com and include the link in your > > > posting. > > > > ***** > > > > > > > > You can also connect a monitor to the RTC to see what it’s doing and > if > > > > it > > > > gives a clue to what to try. > > > > > > > > -Esteban > > > > > > > > > > > > On Sun, Jun 28, 2020 at 9:37 PM Prof. Dr. K. S. Krishnamurthy < > > > > [hidden email]> wrote: > > > > > > > >> ***** > > > >> To join, leave or search the confocal microscopy listserv, go to: > > > >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > > >> Post images on http://www.imgur.com and include the link in your > > > >> posting. > > > >> ***** > > > >> > > > >> Good morning. Thank you for the input. Our computer engineer tried > > using > > > >> the keyboard, but did not succeed in booting. > > > >> > > > > > > > > > > > > > -- > > > Centre for Nano and Soft Matter Sciences > > > PO Box 1329 > > > Jalahalli > > > bangalore 560 013 > > > Ph: +91 80 23084236 (O), Cell:9845284587 > > > > > > > > -- > -- > Zdenek Svindrych, Ph.D. > Research Scientist - Microscopy Imaging Specialist > Department of Biochemistry and Cell Biology > Geisel School of Medicine at Dartmouth > |
|
|
In reply to this post by Prof. Dr. K. S. Krishnamurthy
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** have you tried changing the motherboard battery? we had similar issue a few years ago and that was the problem best rodrigo.- El 28/06/2020 a las 01:52 a. m., Prof. Dr. K. S. Krishnamurthy escribió: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > We have a 2008 model Axioimager, coupled to an LSM5 exciter, used mainly for polarization studies on liquid crystals. Recently, the Zen software failed to launch with "fail to connect to RTC..." message. Resetting did not work. On the monitor, however, 'Real time connection connected..' message appears. It is suspected that the IDE, transcend 128 MB flash module, used to boot may have been corrupted. If a new IDE flash module is purchased, the program data will have to be written into it. Is there a way of getting the data? Since the model is out of service, local Zeiss rep has expressed inability to extend technical support. It seems such a pity that for the sake of the IDE, the instrument should remain defunct. We would be indebted for any suggestions to bring the confocal onroad. |