Login  Register

Far-red filter set

classic Classic list List threaded Threaded
17 messages Options Options
Embed post
Permalink
Kenneth Bowitz Larsen Kenneth Bowitz Larsen
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Far-red filter set

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Far-red filter set

Hi,

we're considering buying a Y5 filter cube from Leica to see the far-red Alexas, DRAQ5 etc. but then someone mentioned to me that many (most?) people are unable to see far-red light. Does anyone know if this is true, or have experience with this (or a similar) filter cube? It has a BP 700/75 suppression filter.


Kenneth

simon walker (BI) simon walker (BI)
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Far-red filter set

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Far-red filter set

Generally the emission of these dyes is considered to be beyond the range of the human eye, but there are papers which describe direct observation of the shorter emission wavelengths, e.g. J Histochem Cytochem. 2000 Mar 48(3):437-44

 

Simon


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Kenneth Bowitz Larsen
Sent: 07 March 2008 11:43
To: [hidden email]
Subject: Far-red filter set

 

Hi,

we're considering buying a Y5 filter cube from Leica to see the far-red Alexas, DRAQ5 etc. but then someone mentioned to me that many (most?) people are unable to see far-red light. Does anyone know if this is true, or have experience with this (or a similar) filter cube? It has a BP 700/75 suppression filter.

 

Kenneth

Peter Humphreys Peter Humphreys
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Far-red filter set

In reply to this post by Kenneth Bowitz Larsen
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Far-red filter set

We have a y5, can see 594, cy5 etc.

Dark red, strong signal more apparent when viewing, haven’t met anyone yet who can’t see it.

 

Peter

 

Imaging Facility Manager

Wellcome Trust Centre for Stem Cell Research

University of Cambridge

Tennis Court Road

Cambridge

CB2 1QR


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Kenneth Bowitz Larsen
Sent: 07 March 2008 11:43
To: [hidden email]
Subject: Far-red filter set

 

Hi,

we're considering buying a Y5 filter cube from Leica to see the far-red Alexas, DRAQ5 etc. but then someone mentioned to me that many (most?) people are unable to see far-red light. Does anyone know if this is true, or have experience with this (or a similar) filter cube? It has a BP 700/75 suppression filter.

 

Kenneth

Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Far-red filter set

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Far-red filter set
We were recently imaging bacteria labelled with Atto 647.  Nobody was
able to see them by eye, though the signal was excellent in confocal,
particularly with APD detection.  This was a Leica filter set - I suppose
it was Y5 but don't know for sure.  But the light source may not have been
ideal for excitation either. 
 
                                                                                           Guy
 

Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net

 


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Peter Humphreys
Sent: Friday, 7 March 2008 11:03 PM
To: [hidden email]
Subject: Re: Far-red filter set

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

We have a y5, can see 594, cy5 etc.

Dark red, strong signal more apparent when viewing, haven’t met anyone yet who can’t see it.

 

Peter

 

Imaging Facility Manager

Wellcome Trust Centre for Stem Cell Research

University of Cambridge

Tennis Court Road

Cambridge

CB2 1QR


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Kenneth Bowitz Larsen
Sent: 07 March 2008 11:43
To: [hidden email]
Subject: Far-red filter set

 

Hi,

we're considering buying a Y5 filter cube from Leica to see the far-red Alexas, DRAQ5 etc. but then someone mentioned to me that many (most?) people are unable to see far-red light. Does anyone know if this is true, or have experience with this (or a similar) filter cube? It has a BP 700/75 suppression filter.

 

Kenneth


No virus found in this incoming message.
Checked by AVG Free Edition.
Version: 7.5.516 / Virus Database: 269.21.4/1309 - Release Date: 3/03/2008 6:50 PM


No virus found in this outgoing message.
Checked by AVG Free Edition.
Version: 7.5.516 / Virus Database: 269.21.4/1309 - Release Date: 3/03/2008 6:50 PM

Schwartz, Owen (NIH/NIAID) [E] Schwartz, Owen (NIH/NIAID) [E]
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Far-red filter set

In reply to this post by Kenneth Bowitz Larsen
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Far-red filter set

I have observed as well that people in general cannot see out into the far red wavelengths.  Sometimes if the sample is very bright, you can see it, but in most cases not.  The far-red cubes do work very well however for most ccd cameras.  Be aware that many cameras have a blue filter installed to reduce IR reaching the ccd.  This might have to be removed (consult manufacturer) in order to image in the far-red.  

 

Cheers,

 

Owen

 

 

Owen M. Schwartz PhD
Head, Biological Imaging Facility
NIAID-NIH
Bldg 4 room B2-30
4 Center Drive
Bethesda, MD 20892
Ph 301-402-5883
Fax 301-480-2882


From: Kenneth Bowitz Larsen [mailto:[hidden email]]
Sent: Friday, March 07, 2008 6:43 AM
To: [hidden email]
Subject: Far-red filter set

 

Hi,

we're considering buying a Y5 filter cube from Leica to see the far-red Alexas, DRAQ5 etc. but then someone mentioned to me that many (most?) people are unable to see far-red light. Does anyone know if this is true, or have experience with this (or a similar) filter cube? It has a BP 700/75 suppression filter.

 

Kenneth

George Ring George Ring
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Far-red filter set

In reply to this post by Kenneth Bowitz Larsen
Far-red filter set
Hi, Ken,
 
SUPPOSEDLY the human eye can see up to 700nm, and women can see longer wavelengths than men.  That said, I agree with the others who responded - I haven't met anyone who can see it.  A camera, however, or confocal can detect far red.
 
Good luck,
 
George
 
George Ring, Ph.D.
Dept. of Cell and Developmental Biology
SUNY Upstate Medical University
750 E. Adams St.
Syracuse NY  13210
Tel. (315) 464-8595
FAX (315) 464-8535
email: [hidden email]

>>> Kenneth Bowitz Larsen <[hidden email]> 3/7/08 6:42:44 AM >>>
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi,

we're considering buying a Y5 filter cube from Leica to see the far-red Alexas, DRAQ5 etc. but then someone mentioned to me that many (most?) people are unable to see far-red light. Does anyone know if this is true, or have experience with this (or a similar) filter cube? It has a BP 700/75 suppression filter.


Kenneth

Hege Avsnes Dale Hege Avsnes Dale
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Fixing GFP positive tissue after freeze sectioning

In reply to this post by Schwartz, Owen (NIH/NIAID) [E]
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi.
Do anyone have a protocol for good preservation of GFP in tissue after
freeze sectioning?
The tissue is a tumor with EGFP-expressing cells and need only a
fixation that will preserve GFP (no permabilization  needed). Would it
also be a better idea to fix before sectioning?

Hege
Martin Wessendorf-2 Martin Wessendorf-2
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Far-red filter set

In reply to this post by George Ring
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

George Ring wrote:

> SUPPOSEDLY the human eye can see up to 700nm, and women can see longer
> wavelengths than men.  That said, I agree with the others who responded
> - I haven't met anyone who can see it.  A camera, however, or
> confocal can detect far red.

I think that the human eye has been shown to be capable of detecting
photons out beyond 1000 nm--it's just the probability of detection gets
pretty low.

--Based on my own lab's experience, I agree with Peter's comment--I
haven't met many people who can't see Cy5, for instance.  However,
intensity is all-important.  Filled cells that've been stained with
Cy5-streptavidin are easy to see with a 20x 0.7 NA lens by eye and
occasionally visible even at 4x.  However, smaller structures with
weaker labeling (e.g. terminals) are much harder to see and even with
brightly labeled cells, the details can get lost.

Re: IR filters on CCD cameras--they indeed can be deadly when trying to
image far-red fluorophores!

Martin
--
Martin Wessendorf, Ph.D.                   office: (612) 626-0145
Assoc Prof, Dept Neuroscience                 lab: (612) 624-2991
University of Minnesota             Preferred FAX: (612) 624-8118
6-145 Jackson Hall, 321 Church St. SE    Dept Fax: (612) 626-5009
Minneapolis, MN  55455
**MY E-MAIL ADDRESS HAS CHANGED.  PLEASE USE [hidden email] **
Michael Herron Michael Herron
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Far-red filter set

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal A nice paper on it here for wide field:  http://www.jhc.org/cgi/reprint/48/3/437

The Jackson website discusses Cy5 a bit here: http://www.jacksonimmuno.com/technical/f-cy3-5.asp

On Mar 7, 2008, at 9:27 AM, Martin Wessendorf wrote:

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

George Ring wrote:

SUPPOSEDLY the human eye can see up to 700nm, and women can see longer wavelengths than men.  That said, I agree with the others who responded - I haven't met anyone who can see it.  A camera, however, or confocal can detect far red.

I think that the human eye has been shown to be capable of detecting photons out beyond 1000 nm--it's just the probability of detection gets pretty low.

--Based on my own lab's experience, I agree with Peter's comment--I haven't met many people who can't see Cy5, for instance.  However, intensity is all-important.  Filled cells that've been stained with Cy5-streptavidin are easy to see with a 20x 0.7 NA lens by eye and occasionally visible even at 4x.  However, smaller structures with weaker labeling (e.g. terminals) are much harder to see and even with brightly labeled cells, the details can get lost.

Re: IR filters on CCD cameras--they indeed can be deadly when trying to image far-red fluorophores!

Martin
-- 
Martin Wessendorf, Ph.D.                   office: (612) 626-0145
Assoc Prof, Dept Neuroscience                 lab: (612) 624-2991
University of Minnesota             Preferred FAX: (612) 624-8118
6-145 Jackson Hall, 321 Church St. SE    Dept Fax: (612) 626-5009
Minneapolis, MN  55455
**MY E-MAIL ADDRESS HAS CHANGED.  PLEASE USE [hidden email] **

Michael J. Herron,  U of MN, Dept. of Entomology
     612-624-3688 (office) 612-625-5299 (FAX)



Jennifer Waters Jennifer Waters
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Far-red filter set

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Here's my guess:
While the peak intensity of Cy5 is not visible, it starts emitting (if you look at the spectra) around 620nm which most of us can see.  So if there is a high enough concentration of probe, the small percentage of photons that emit from Cy5 in the lower wavelengths will be detectable.
Jennifer



On Fri, Mar 7, 2008 at 11:11 AM, Michael Herron <[hidden email]> wrote:
A nice paper on it here for wide field:  http://www.jhc.org/cgi/reprint/48/3/437

The Jackson website discusses Cy5 a bit here: http://www.jacksonimmuno.com/technical/f-cy3-5.asp

On Mar 7, 2008, at 9:27 AM, Martin Wessendorf wrote:

Search the CONFOCAL archive at

George Ring wrote:

SUPPOSEDLY the human eye can see up to 700nm, and women can see longer wavelengths than men.  That said, I agree with the others who responded - I haven't met anyone who can see it.  A camera, however, or confocal can detect far red.

I think that the human eye has been shown to be capable of detecting photons out beyond 1000 nm--it's just the probability of detection gets pretty low.

--Based on my own lab's experience, I agree with Peter's comment--I haven't met many people who can't see Cy5, for instance.  However, intensity is all-important.  Filled cells that've been stained with Cy5-streptavidin are easy to see with a 20x 0.7 NA lens by eye and occasionally visible even at 4x.  However, smaller structures with weaker labeling (e.g. terminals) are much harder to see and even with brightly labeled cells, the details can get lost.

Re: IR filters on CCD cameras--they indeed can be deadly when trying to image far-red fluorophores!

Martin
-- 
Martin Wessendorf, Ph.D.                   office: (612) 626-0145
Assoc Prof, Dept Neuroscience                 lab: (612) 624-2991
University of Minnesota             Preferred FAX: (612) 624-8118
6-145 Jackson Hall, 321 Church St. SE    Dept Fax: (612) 626-5009
Minneapolis, MN  55455
**MY E-MAIL ADDRESS HAS CHANGED.  PLEASE USE [hidden email] **

Michael J. Herron,  U of MN, Dept. of Entomology
     612-624-3688 (office) 612-625-5299 (FAX)






--
Jennifer Waters, Ph.D.
Director, Nikon Imaging Center at Harvard Medical School
Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Far-red filter set

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Well, that depends where you are exciting.  I guess if you excite in the yellow
that could work but exciting at 640 rather rules it out.  I'm sure it is just a
sensitivity issue - I can see 750nm laser light easily enough.  And if you
have a cell layer or similar there will be enough glow around to find things.
But little bacteria are difficult.  We had 3 young men, I old man (me) and
1 young woman try and none of us could see them.   We had to find them
by phase - even that wasn't so easy since we'd worked hard to match
the refractive index!
 
                                                                                           Guy

Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net

 


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Jennifer Waters
Sent: Saturday, 8 March 2008 5:07 AM
To: [hidden email]
Subject: Re: Far-red filter set

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Here's my guess:
While the peak intensity of Cy5 is not visible, it starts emitting (if you look at the spectra) around 620nm which most of us can see.  So if there is a high enough concentration of probe, the small percentage of photons that emit from Cy5 in the lower wavelengths will be detectable.
Jennifer



On Fri, Mar 7, 2008 at 11:11 AM, Michael Herron <[hidden email]> wrote:
A nice paper on it here for wide field:  http://www.jhc.org/cgi/reprint/48/3/437

The Jackson website discusses Cy5 a bit here: http://www.jacksonimmuno.com/technical/f-cy3-5.asp

On Mar 7, 2008, at 9:27 AM, Martin Wessendorf wrote:

Search the CONFOCAL archive at

George Ring wrote:

SUPPOSEDLY the human eye can see up to 700nm, and women can see longer wavelengths than men.  That said, I agree with the others who responded - I haven't met anyone who can see it.  A camera, however, or confocal can detect far red.

I think that the human eye has been shown to be capable of detecting photons out beyond 1000 nm--it's just the probability of detection gets pretty low.

--Based on my own lab's experience, I agree with Peter's comment--I haven't met many people who can't see Cy5, for instance.  However, intensity is all-important.  Filled cells that've been stained with Cy5-streptavidin are easy to see with a 20x 0.7 NA lens by eye and occasionally visible even at 4x.  However, smaller structures with weaker labeling (e.g. terminals) are much harder to see and even with brightly labeled cells, the details can get lost.

Re: IR filters on CCD cameras--they indeed can be deadly when trying to image far-red fluorophores!

Martin
-- 
Martin Wessendorf, Ph.D.                   office: (612) 626-0145
Assoc Prof, Dept Neuroscience                 lab: (612) 624-2991
University of Minnesota             Preferred FAX: (612) 624-8118
6-145 Jackson Hall, 321 Church St. SE    Dept Fax: (612) 626-5009
Minneapolis, MN  55455
**MY E-MAIL ADDRESS HAS CHANGED.  PLEASE USE [hidden email] **

Michael J. Herron,  U of MN, Dept. of Entomology
     612-624-3688 (office) 612-625-5299 (FAX)






--
Jennifer Waters, Ph.D.
Director, Nikon Imaging Center at Harvard Medical School

No virus found in this incoming message.
Checked by AVG Free Edition.
Version: 7.5.516 / Virus Database: 269.21.4/1309 - Release Date: 3/03/2008 6:50 PM


No virus found in this outgoing message.
Checked by AVG Free Edition.
Version: 7.5.516 / Virus Database: 269.21.4/1309 - Release Date: 3/03/2008 6:50 PM

Mayandi Sivaguru Mayandi Sivaguru
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Far-red filter set

In reply to this post by Kenneth Bowitz Larsen
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Kenneth, in my experience, using a M200, D640/20 (Ex), D680/30 (Em), 660 DCLP (BS), a spot size of around 10 micron diameter having an 8 bit intensity of over 200 is clearly visible to my eyes as well as to our graduate student. But   when you project this image to MRm camera (1388x1040) we typically set a exposure of 30-50 ms (which yield spot intensity of over 200 in 8 bits).
Shiv
 

At 05:42 AM 3/7/2008, you wrote:
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi,

we're considering buying a Y5 filter cube from Leica to see the far-red Alexas, DRAQ5 etc. but then someone mentioned to me that many (most?) people are unable to see far-red light. Does anyone know if this is true, or have experience with this (or a similar) filter cube? It has a BP 700/75 suppression filter.

Kenneth

Mayandi Sivaguru, PhD, PhD
Microscopy Facility Manager
8, Institute for Genomic Biology
University of Illinois at Urbana-Champaign
1206 West Gregory Dr.
Urbana, IL 61801 USA

Office: 217.333.1214
Fax: 217.244.2496
[hidden email]
http://core.igb.uiuc.edu

Kirill Ukhanov Kirill Ukhanov
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Fixing GFP positive tissue after freeze sectioning

In reply to this post by Hege Avsnes Dale
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear Hege,

GFP could be quenched by aldehyde fixation, on the other hand you need to
fix the tissue to avoid wash out of GFP as it is mostly located in cytoplasm
(well, depending how its expression is controlled). Poor fixation may result
in a loss of the GFP from the tissue even if you think of not using
permeabilization etc. I would recommend fixing your tissue before
cryosectioning and then use anti-GFP antibody to enhance native GFP (if you
find the signal too weak). I remember I found an extensive discussion on
this issue here in the list or elsewhere in microscopy forums.

good luck

Kirill Ukhanov
University of Florida

On Fri, 7 Mar 2008 16:22:06 +0100, Hege Avsnes Dale
<[hidden email]> wrote:

>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>Hi.
>Do anyone have a protocol for good preservation of GFP in tissue after
>freeze sectioning?
>The tissue is a tumor with EGFP-expressing cells and need only a
>fixation that will preserve GFP (no permabilization  needed). Would it
>also be a better idea to fix before sectioning?
>
>Hege
>=========================================================================
Kirill Ukhanov Kirill Ukhanov
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Fixing GFP positive tissue after freeze sectioning

In reply to this post by Hege Avsnes Dale
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear Hege,

GFP could be quenched by aldehyde fixation, on the other hand you need to
fix the tissue to avoid wash out of GFP as it is mostly located in cytoplasm
(well, depending how its expression is controlled). Poor fixation may result
in a loss of the GFP from the tissue even if you think of not using
permeabilization etc. I would recommend fixing your tissue before
cryosectioning and then use anti-GFP antibody to enhance native GFP (if you
find the signal too weak). I remember I found an extensive discussion on
this issue here in the list or elsewhere in microscopy forums.

good luck

Kirill Ukhanov
University of Florida


On Fri, 7 Mar 2008 16:22:06 +0100, Hege Avsnes Dale
<[hidden email]> wrote:

>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>Hi.
>Do anyone have a protocol for good preservation of GFP in tissue after
>freeze sectioning?
>The tissue is a tumor with EGFP-expressing cells and need only a
>fixation that will preserve GFP (no permabilization  needed). Would it
>also be a better idea to fix before sectioning?
>
>Hege
>=========================================================================
Michael Weber-4 Michael Weber-4
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Far-red filter set

In reply to this post by Kenneth Bowitz Larsen
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Kenneth,

does the Leica Y5 set consist of hard-coated filters and dichromatic
mirrors? I recommend to go for this type of filters (higher
transmission, less backreflection). Chroma and Semrock do offer them.

Michael


Kenneth Bowitz Larsen wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi,
>
> we're considering buying a Y5 filter cube from Leica to see the far-red
> Alexas, DRAQ5 etc. but then someone mentioned to me that many (most?)
> people are unable to see far-red light. Does anyone know if this is
> true, or have experience with this (or a similar) filter cube? It has a
> BP 700/75 suppression filter.
>
>
> Kenneth
Caroline Bass Caroline Bass
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Fixing GFP positive tissue after freeze sectioning

In reply to this post by Kirill Ukhanov
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I regularly fix brain tissue that is expressing GFP (via injection of a
viral vector).  Standard NBF works fine and I have no problems.  I looked
into this issue extensively a few years ago, and I can say that two things
killed my signal: not fixing the tissue, and allowing fresh tissue to
dehydrate.

I have fixed my brains in two ways, perfusion of the animal and immersion in
NBF.  Since the brain (mouse) is a large organ I cut it in three parts and
just dropped it in for a quick and easy fix.  This worked fine, but I only
used it when I needed to get a quick look at the expression zone.


On 3/10/08 9:34 AM, "Kirill Ukhanov" <[hidden email]> wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Dear Hege,
>
> GFP could be quenched by aldehyde fixation, on the other hand you need to
> fix the tissue to avoid wash out of GFP as it is mostly located in cytoplasm
> (well, depending how its expression is controlled). Poor fixation may result
> in a loss of the GFP from the tissue even if you think of not using
> permeabilization etc. I would recommend fixing your tissue before
> cryosectioning and then use anti-GFP antibody to enhance native GFP (if you
> find the signal too weak). I remember I found an extensive discussion on
> this issue here in the list or elsewhere in microscopy forums.
>
> good luck
>
> Kirill Ukhanov
> University of Florida
>
>
> On Fri, 7 Mar 2008 16:22:06 +0100, Hege Avsnes Dale
> <[hidden email]> wrote:
>
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>> Hi.
>> Do anyone have a protocol for good preservation of GFP in tissue after
>> freeze sectioning?
>> The tissue is a tumor with EGFP-expressing cells and need only a
>> fixation that will preserve GFP (no permabilization  needed). Would it
>> also be a better idea to fix before sectioning?
>>
>> Hege
>> =========================================================================
Melissa Gonzalez Melissa Gonzalez
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Fixing GFP positive tissue after freeze sectioning

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Can you directly observe GFP in fresh frozen tissues?
I always fix in 4% PFA, cryoprotect, and then freeze. Those sections
always look great.

But any time I have received fresh frozens they turn up negative. Post
fixing in PFA doesn't work either.

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Caroline Bass
Sent: Monday, March 10, 2008 6:55 AM
To: [hidden email]
Subject: Re: Fixing GFP positive tissue after freeze sectioning

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I regularly fix brain tissue that is expressing GFP (via injection of a
viral vector).  Standard NBF works fine and I have no problems.  I
looked
into this issue extensively a few years ago, and I can say that two
things
killed my signal: not fixing the tissue, and allowing fresh tissue to
dehydrate.

I have fixed my brains in two ways, perfusion of the animal and
immersion in
NBF.  Since the brain (mouse) is a large organ I cut it in three parts
and
just dropped it in for a quick and easy fix.  This worked fine, but I
only
used it when I needed to get a quick look at the expression zone.


On 3/10/08 9:34 AM, "Kirill Ukhanov" <[hidden email]> wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Dear Hege,
>
> GFP could be quenched by aldehyde fixation, on the other hand you need
to
> fix the tissue to avoid wash out of GFP as it is mostly located in
cytoplasm
> (well, depending how its expression is controlled). Poor fixation may
result
> in a loss of the GFP from the tissue even if you think of not using
> permeabilization etc. I would recommend fixing your tissue before
> cryosectioning and then use anti-GFP antibody to enhance native GFP
(if you
> find the signal too weak). I remember I found an extensive discussion
on

> this issue here in the list or elsewhere in microscopy forums.
>
> good luck
>
> Kirill Ukhanov
> University of Florida
>
>
> On Fri, 7 Mar 2008 16:22:06 +0100, Hege Avsnes Dale
> <[hidden email]> wrote:
>
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>> Hi.
>> Do anyone have a protocol for good preservation of GFP in tissue
after
>> freeze sectioning?
>> The tissue is a tumor with EGFP-expressing cells and need only a
>> fixation that will preserve GFP (no permabilization  needed). Would
it
>> also be a better idea to fix before sectioning?
>>
>> Hege
>>
========================================================================
=