Login  Register

First Contact on Mirrors and Filters

classic Classic list List threaded Threaded
3 messages Options Options
Embed post
Permalink
Claire Brown Claire Brown
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

First Contact on Mirrors and Filters

I wonder if anyone has used first contact on dichroic mirrors, excitation or
emission filters? We have both soft and hard coated surfaces. I would assume
it would be okay for hard coated filters but I am reluctant to try it on
soft coated ones.

Thanks for any thoughts on this.

Sincerely,

Claire
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: First Contact on Mirrors and Filters

Apparently you can use the stuff on diffraction gratings, so I'd
assume it would be ok.  I'd get a small sample and try a tiny daub of
it on the very corner of the filter to see what happens.

Craig


On Mon, May 25, 2009 at 6:22 PM, Claire Brown <[hidden email]> wrote:

> I wonder if anyone has used first contact on dichroic mirrors, excitation or
> emission filters? We have both soft and hard coated surfaces. I would assume
> it would be okay for hard coated filters but I am reluctant to try it on
> soft coated ones.
>
> Thanks for any thoughts on this.
>
> Sincerely,
>
> Claire
>
Claire Brown Claire Brown
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: First Contact on Mirrors and Filters

In reply to this post by Claire Brown
I heard back from Chroma Technologies that it is safe to use on their
mirrors and filters. They did say it works well on some residues but not
others, I believe it does not completely take care of finger prints.