Ricardo Henriques |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear colleagues, I wanted to do a small highlight on a preprint we've uploaded to bioRxiv yesterday: "Fix your membrane receptor imaging: Actin cytoskeleton and CD4 membrane organization disruption by chemical fixation" https://www.biorxiv.org/content/early/2018/10/23/450635 With Super-Resolution we show careful fixation is needed to correctly represent the actin cortex and membrane receptor topology. Suboptimal fixation easily leads to artefacts. Hope this may be of interest to you. All the best, -Ricardo -- Ricardo Henriques, Associate Professor University College London and The Francis Crick Institute [https://docs.google.com/uc?export=download&id=0BzbBSUuAJup_NDRXcWM0T1VpQXc&revid=0BzbBSUuAJup_ZzVnRS9EK3FwTUNFbnVhdmx5Y1Azbjl4NytZPQ]<http://www.ucl.ac.uk/lmcb/research-group/ricardo-henriques-research-group> MRC-Laboratory for Molecular Cell Biology University College London Gower Street London WC1E 6BT United Kingdom Twitter: @HenriquesLab<https://twitter.com/HenriquesLab> ORCID: https://orcid.org/0000-0002-2043-5234<https://orcid.org/0000-0002-2043-5234/print> Google Scholar: https://scholar.google.co.uk/citations?user=-peQ4ZsAAAAJ&hl=en |
|
Cammer, Michael |
|
JAMES B PAWLEY |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Thanks for the reminder Ricardo. If you aim at electron microscope resolution on fixed specimens, you need to use electron microscope preparative procedures. This is why we included Chapter 18 in The Handbook. GUIDING PRINCIPLES OF SPECIMEN PRESERVATION FOR CONFOCAL FLUORESCENCE MICROSCOPY by Robert Bacallao, Sadaf Sohrab, and Carrie Phillips Perhaps others may find this chapter useful. Best, Jim Pawley James and Christine Pawley, 5446 Burley Place, Box 2348, Sechelt BC, Canada, V0N3A0 [hidden email]<mailto:[hidden email]>, Phone 1-604-885-0840, cell 1-604-989-6146 On Oct 24, 2018, at 12:33 PM, Cammer, Michael <[hidden email]<mailto:[hidden email]>> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Perhaps of interest: http://microscopynotes.com/IRM/index.html -----Original Message----- From: Confocal Microscopy List <[hidden email]> On Behalf Of Henriques, Ricardo Sent: Wednesday, October 24, 2018 1:57 PM To: [hidden email] Subject: Fixation artefacts when imaging actin cortex and transmembrane proteins by Super-Resolution Dear colleagues, I wanted to do a small highlight on a preprint we've uploaded to bioRxiv yesterday: "Fix your membrane receptor imaging: Actin cytoskeleton and CD4 membrane organization disruption by chemical fixation" https://urldefense.proofpoint.com/v2/url?u=https-3A__www.biorxiv.org_content_early_2018_10_23_450635&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=rR_TAsuHCQvOR7r44A0KnTFHHb4TJ2AwzNGzpKasbzg&s=mWNqUGoeqFzAXCeVNVCIuV-VgJfkt_oi5Oj-qjTX-HQ&e= With Super-Resolution we show careful fixation is needed to correctly represent the actin cortex and membrane receptor topology. Suboptimal fixation easily leads to artefacts. Hope this may be of interest to you. All the best, -Ricardo -- Ricardo Henriques, Associate Professor University College London and The Francis Crick Institute [https://urldefense.proofpoint.com/v2/url?u=https-3A__docs.google.com_uc-3Fexport-3Ddownload-26id-3D0BzbBSUuAJup-5FNDRXcWM0T1VpQXc-26revid-3D0BzbBSUuAJup-5FZzVnRS9EK3FwTUNFbnVhdmx5Y1Azbjl4NytZPQ&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=rR_TAsuHCQvOR7r44A0KnTFHHb4TJ2AwzNGzpKasbzg&s=2ghNHJsNSekUKC7VvcBxFwUJU6Nx-K3hz88BbmSwmTM&e=]<https://urldefense.proofpoint.com/v2/url?u=http-3A__www.ucl.ac.uk_lmcb_research-2Dgroup_ricardo-2Dhenriques-2Dresearch-2Dgroup&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=rR_TAsuHCQvOR7r44A0KnTFHHb4TJ2AwzNGzpKasbzg&s=NIrZnYPglsq5_eN8_PwptRP0o2GRveQ8T7ZjAMhY2bg&e=> MRC-Laboratory for Molecular Cell Biology University College London Gower Street London WC1E 6BT United Kingdom Twitter: @HenriquesLab<https://urldefense.proofpoint.com/v2/url?u=https-3A__twitter.com_HenriquesLab&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=rR_TAsuHCQvOR7r44A0KnTFHHb4TJ2AwzNGzpKasbzg&s=wnVufIovrOOX9vlVRVQVXeteG_NeJ2dlOlKaqrLmFjM&e=> ORCID: https://urldefense.proofpoint.com/v2/url?u=https-3A__orcid.org_0000-2D0002-2D2043-2D5234&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=rR_TAsuHCQvOR7r44A0KnTFHHb4TJ2AwzNGzpKasbzg&s=W-j-Zbz3To3v38rmRWcc6wgZ_8YeWuw05tdB6zVZKm4&e=<https://urldefense.proofpoint.com/v2/url?u=https-3A__orcid.org_0000-2D0002-2D2043-2D5234_print&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=rR_TAsuHCQvOR7r44A0KnTFHHb4TJ2AwzNGzpKasbzg&s=5c_uGFkk1I9y2LuS1cHBWwUjH7eAk9sXysg4lCX2l-U&e=> Google Scholar: https://urldefense.proofpoint.com/v2/url?u=https-3A__scholar.google.co.uk_citations-3Fuser-3D-2DpeQ4ZsAAAAJ-26hl-3Den&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=rR_TAsuHCQvOR7r44A0KnTFHHb4TJ2AwzNGzpKasbzg&s=VOmHzLKr3Xv3NJmre7fEUUjxqmAnAOOXw7FTg6C0hmU&e= ------------------------------------------------------------ This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email. ================================= |
|
George McNamara |
|
|
In reply to this post by Cammer, Michael
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Speaking of IRM and general reflection, see https://www.rms.org.uk/study-read/infocus-magazine/infocus-listing/reflection-contrast-microscopy-review.html 1 September 2017Issue 47 PLOEM J.S. and PRINS F.A. Reflection-Contrast Microscopy - Review This article gives an overview of the applications of Reflection-Contrast Microscopy (RCM). DOI: 10.22443/rms.inf.1.152 A short explanation of the antiflex and oblique illumination methods, used with RCM is presented. The advantage of using ultrathin sections for light microscopy with RCM is discussed. Further details can also be found on the website: www.ploem-reflection-contrast-microscopy.com. A video illustrating RCM is available on YouTube: www.youtube.com/watch?v=sNQGu3nDPX8 *** All RMS InFocus articles are open access after 1 year. Same issue has light sheet review and an earlier issue has Raspberry Pi. enjoy, George p.s. confocal IRM HeLa https://works.bepress.com/gmcnamara/10/ On 10/24/2018 3:33 PM, Cammer, Michael wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Perhaps of interest: http://microscopynotes.com/IRM/index.html > > -----Original Message----- > From: Confocal Microscopy List <[hidden email]> On Behalf Of Henriques, Ricardo > Sent: Wednesday, October 24, 2018 1:57 PM > To: [hidden email] > Subject: Fixation artefacts when imaging actin cortex and transmembrane proteins by Super-Resolution > > > Dear colleagues, > > I wanted to do a small highlight on a preprint we've uploaded to bioRxiv yesterday: > > "Fix your membrane receptor imaging: Actin cytoskeleton and CD4 membrane organization disruption by chemical fixation" > > https://urldefense.proofpoint.com/v2/url?u=https-3A__www.biorxiv.org_content_early_2018_10_23_450635&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=rR_TAsuHCQvOR7r44A0KnTFHHb4TJ2AwzNGzpKasbzg&s=mWNqUGoeqFzAXCeVNVCIuV-VgJfkt_oi5Oj-qjTX-HQ&e= > > > > With Super-Resolution we show careful fixation is needed to correctly represent the actin cortex and membrane receptor topology. Suboptimal fixation easily leads to artefacts. Hope this may be of interest to you. > > > > All the best, > > -Ricardo > > > > -- > > Ricardo Henriques, Associate Professor > > University College London and The Francis Crick Institute > > > > [https://urldefense.proofpoint.com/v2/url?u=https-3A__docs.google.com_uc-3Fexport-3Ddownload-26id-3D0BzbBSUuAJup-5FNDRXcWM0T1VpQXc-26revid-3D0BzbBSUuAJup-5FZzVnRS9EK3FwTUNFbnVhdmx5Y1Azbjl4NytZPQ&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=rR_TAsuHCQvOR7r44A0KnTFHHb4TJ2AwzNGzpKasbzg&s=2ghNHJsNSekUKC7VvcBxFwUJU6Nx-K3hz88BbmSwmTM&e=]<https://urldefense.proofpoint.com/v2/url?u=http-3A__www.ucl.ac.uk_lmcb_research-2Dgroup_ricardo-2Dhenriques-2Dresearch-2Dgroup&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=rR_TAsuHCQvOR7r44A0KnTFHHb4TJ2AwzNGzpKasbzg&s=NIrZnYPglsq5_eN8_PwptRP0o2GRveQ8T7ZjAMhY2bg&e=> > > MRC-Laboratory for Molecular Cell Biology > > University College London > > Gower Street > > London WC1E 6BT > > United Kingdom > > > > Twitter: @HenriquesLab<https://urldefense.proofpoint.com/v2/url?u=https-3A__twitter.com_HenriquesLab&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=rR_TAsuHCQvOR7r44A0KnTFHHb4TJ2AwzNGzpKasbzg&s=wnVufIovrOOX9vlVRVQVXeteG_NeJ2dlOlKaqrLmFjM&e=> > > ORCID: https://urldefense.proofpoint.com/v2/url?u=https-3A__orcid.org_0000-2D0002-2D2043-2D5234&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=rR_TAsuHCQvOR7r44A0KnTFHHb4TJ2AwzNGzpKasbzg&s=W-j-Zbz3To3v38rmRWcc6wgZ_8YeWuw05tdB6zVZKm4&e=<https://urldefense.proofpoint.com/v2/url?u=https-3A__orcid.org_0000-2D0002-2D2043-2D5234_print&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=rR_TAsuHCQvOR7r44A0KnTFHHb4TJ2AwzNGzpKasbzg&s=5c_uGFkk1I9y2LuS1cHBWwUjH7eAk9sXysg4lCX2l-U&e=> > > Google Scholar: https://urldefense.proofpoint.com/v2/url?u=https-3A__scholar.google.co.uk_citations-3Fuser-3D-2DpeQ4ZsAAAAJ-26hl-3Den&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=rR_TAsuHCQvOR7r44A0KnTFHHb4TJ2AwzNGzpKasbzg&s=VOmHzLKr3Xv3NJmre7fEUUjxqmAnAOOXw7FTg6C0hmU&e= > > > ------------------------------------------------------------ > This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email. > ================================= > |
|
Ingela Parmryd-3 |
| Free forum by Nabble | Edit this page |