Jacqueline Ross |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear All, Does anyone know of a company who can supply fluorescent spheres in this size range suitable for confocal imaging? I know people are usually looking for standards at the other end of the scale! I've had a look at Bangs Laboratories and Spherotech but they don't seem to have anything large enough... Kind regards, Jacqui Jacqueline Ross Biomedical Imaging Microscopist Biomedical Imaging Research Unit School of Medical Sciences Faculty of Medical & Health Sciences The University of Auckland Private Bag 92019 Auckland 1142, NEW ZEALAND Tel: 64 9 923 7438 Fax: 64 9 373 7484 http://www.fmhs.auckland.ac.nz/sms/biru/ |
|
Franziska Curdt-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** You find them at Thermo Fisher (Invitrogen). E.g. Fluospheres 20nm-15um. Kind regards, Franziska Curdt On Wed, Apr 20, 2016 at 6:36 AM, Jacqui Ross <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear All, > > Does anyone know of a company who can supply fluorescent spheres in this > size range suitable for confocal imaging? I know people are usually looking > for standards at the other end of the scale! > > I've had a look at Bangs Laboratories and Spherotech but they don't seem > to have anything large enough... > > Kind regards, > > Jacqui > > Jacqueline Ross > Biomedical Imaging Microscopist > Biomedical Imaging Research Unit > School of Medical Sciences > Faculty of Medical & Health Sciences > The University of Auckland > Private Bag 92019 > Auckland 1142, NEW ZEALAND > > Tel: 64 9 923 7438 > Fax: 64 9 373 7484 > > http://www.fmhs.auckland.ac.nz/sms/biru/ > |
|
Eric Girard |
|
|
In reply to this post by Jacqueline Ross
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Jacqui, Can you mold agar balls with dye, or oil dye plastic balls or silica spacers? Regards, Eric Girard Sales Engineering / Ingénierie Ventes GiGa Concept Inc Montreal, Quebec, Canada Cell: 514-826-1181 Tel: 514-745-8877 [hidden email] www.gigaconcept.com -----Message d'origine----- De : Confocal Microscopy List [mailto:[hidden email]] De la part de Jacqui Ross Envoyé : April-20-16 12:36 AM À : [hidden email] Objet : Fluorescent spheres - range 100um -1000um ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear All, Does anyone know of a company who can supply fluorescent spheres in this size range suitable for confocal imaging? I know people are usually looking for standards at the other end of the scale! I've had a look at Bangs Laboratories and Spherotech but they don't seem to have anything large enough... Kind regards, Jacqui Jacqueline Ross Biomedical Imaging Microscopist Biomedical Imaging Research Unit School of Medical Sciences Faculty of Medical & Health Sciences The University of Auckland Private Bag 92019 Auckland 1142, NEW ZEALAND Tel: 64 9 923 7438 Fax: 64 9 373 7484 http://www.fmhs.auckland.ac.nz/sms/biru/ |
|
Knecht, David |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** We have gotten many spheres from Discovery Sciences (now Advance Scientific) http://www.advancescientific-usa.com/microparticles Dr. David Knecht Professor of Molecular and Cell Biology Core Microscopy Facility Director University of Connecticut 91 N. Eagleville Rd. Storrs, CT 06269 860-486-2200 On Apr 20, 2016, at 4:25 AM, Eric Girard <[hidden email]<mailto:[hidden email]>> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Jacqui, Can you mold agar balls with dye, or oil dye plastic balls or silica spacers? Regards, Eric Girard Sales Engineering / Ingénierie Ventes GiGa Concept Inc Montreal, Quebec, Canada Cell: 514-826-1181 Tel: 514-745-8877 [hidden email] www.gigaconcept.com -----Message d'origine----- De : Confocal Microscopy List [mailto:[hidden email]] De la part de Jacqui Ross Envoyé : April-20-16 12:36 AM À : [hidden email] Objet : Fluorescent spheres - range 100um -1000um ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear All, Does anyone know of a company who can supply fluorescent spheres in this size range suitable for confocal imaging? I know people are usually looking for standards at the other end of the scale! I've had a look at Bangs Laboratories and Spherotech but they don't seem to have anything large enough... Kind regards, Jacqui Jacqueline Ross Biomedical Imaging Microscopist Biomedical Imaging Research Unit School of Medical Sciences Faculty of Medical & Health Sciences The University of Auckland Private Bag 92019 Auckland 1142, NEW ZEALAND Tel: 64 9 923 7438 Fax: 64 9 373 7484 http://www.fmhs.auckland.ac.nz/sms/biru/ |
|
Steffen Dietzel |
|
|
In reply to this post by Jacqueline Ross
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** You want to check the FWHM of the PSF of your telescope? Ah, no, you write for confocal imaging. Then I really wonder what you plan to do with them. Cucurbita (pumpkin, squash, zucchini) pollen is said to be ~170-180 µm diameter, apparently the largest pollen known. If they are as fluorescent as others, they might do the trick, depending on what you want to do with them. Certainly would come cheaper than the latex spheres. Steffen Am 20.04.2016 um 06:36 schrieb Jacqui Ross: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear All, > > Does anyone know of a company who can supply fluorescent spheres in this size range suitable for confocal imaging? I know people are usually looking for standards at the other end of the scale! > > I've had a look at Bangs Laboratories and Spherotech but they don't seem to have anything large enough... > > Kind regards, > > Jacqui > > Jacqueline Ross > Biomedical Imaging Microscopist > Biomedical Imaging Research Unit > School of Medical Sciences > Faculty of Medical & Health Sciences > The University of Auckland > Private Bag 92019 > Auckland 1142, NEW ZEALAND > > Tel: 64 9 923 7438 > Fax: 64 9 373 7484 > > http://www.fmhs.auckland.ac.nz/sms/biru/ > > > -- > ------------------------------------------------------------ > Steffen Dietzel, PD Dr. rer. nat > Ludwig-Maximilians-Universität München > Biomedical Center (BMC) > Head of the Core Facility Bioimaging > > Großhaderner Straße 9 > D-82152 Planegg-Martinsried > Germany > > http://www.bioimaging.bmc.med.uni-muenchen.de |
|
Jacqueline Ross |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Thanks very much to everyone who replied to my email about fluorescent spheres, I now have a number of different avenues (and techniques) to try. I may even head down to the community garden and see if we still have any pumpkin/zucchini pollen available as suggested by Steffen:-). Cheers, Jacqui Jacqueline Ross Biomedical Imaging Microscopist Biomedical Imaging Research Unit School of Medical Sciences Faculty of Medical & Health Sciences The University of Auckland Private Bag 92019 Auckland 1142, NEW ZEALAND Tel: 64 9 923 7438 Fax: 64 9 373 7484 http://www.fmhs.auckland.ac.nz/sms/biru/ -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Steffen Dietzel Sent: Thursday, 21 April 2016 2:21 a.m. To: [hidden email] Subject: Re: Fluorescent spheres - range 100um -1000um ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** You want to check the FWHM of the PSF of your telescope? Ah, no, you write for confocal imaging. Then I really wonder what you plan to do with them. Cucurbita (pumpkin, squash, zucchini) pollen is said to be ~170-180 µm diameter, apparently the largest pollen known. If they are as fluorescent as others, they might do the trick, depending on what you want to do with them. Certainly would come cheaper than the latex spheres. Steffen Am 20.04.2016 um 06:36 schrieb Jacqui Ross: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear All, > > Does anyone know of a company who can supply fluorescent spheres in this size range suitable for confocal imaging? I know people are usually looking for standards at the other end of the scale! > > I've had a look at Bangs Laboratories and Spherotech but they don't seem to have anything large enough... > > Kind regards, > > Jacqui > > Jacqueline Ross > Biomedical Imaging Microscopist > Biomedical Imaging Research Unit > School of Medical Sciences > Faculty of Medical & Health Sciences > The University of Auckland > Private Bag 92019 > Auckland 1142, NEW ZEALAND > > Tel: 64 9 923 7438 > Fax: 64 9 373 7484 > > http://www.fmhs.auckland.ac.nz/sms/biru/ > > > -- > ------------------------------------------------------------ > Steffen Dietzel, PD Dr. rer. nat > Ludwig-Maximilians-Universität München > Biomedical Center (BMC) > Head of the Core Facility Bioimaging > > Großhaderner Straße 9 > D-82152 Planegg-Martinsried > Germany > > http://www.bioimaging.bmc.med.uni-muenchen.de |