Eric Shelden |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** All: I am looking for a way to fluorescently counter-stain the volume of cultured cells, with a result similar to what would be achieved by having them express GFP in the cytoplasm. I have seen some references to staining with eosin Y. Does anyone have other suggestions? Thanks in advance. Eric |
![]() ![]() |
Cromey, Douglas W - (dcromey) |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Invitrogen's Live/Dead assay includes calcein, a dye that is only taken up by live cells and it only becomes fluorescent when it is exposed to cytoplasmic esterases. It's a green dye and I am pretty sure that you can purchase it separately and not just as part of the kit. Doug ^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^ Douglas W. Cromey, M.S. - Associate Scientific Investigator Dept. of Cellular & Molecular Medicine, University of Arizona 1501 N. Campbell Ave, Tucson, AZ 85724-5044 USA office: AHSC 4212 email: [hidden email] voice: 520-626-2824 fax: 520-626-2097 http://swehsc.pharmacy.arizona.edu/micro Home of: "Microscopy and Imaging Resources on the WWW" UA Microscopy Alliance - http://microscopy.arizona.edu -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Eric Shelden Sent: Wednesday, September 17, 2014 10:49 AM To: [hidden email] Subject: General fluorescent stain for cytoplasm ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** All: I am looking for a way to fluorescently counter-stain the volume of cultured cells, with a result similar to what would be achieved by having them express GFP in the cytoplasm. I have seen some references to staining with eosin Y. Does anyone have other suggestions? Thanks in advance. Eric |
![]() ![]() |
Chew, Teng-Leong |
![]() |
In reply to this post by Eric Shelden
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Invitrogen: CellTracer (several colors to choose from). Best, Leong -- Teng-Leong Chew, PhD Director, Advanced Imaging Center Howard Hughes Medical Institute at Janelia Ashburn, VA 20147 On 9/17/14, 1:49 PM, "Eric Shelden" <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >Post images on http://www.imgur.com and include the link in your posting. >***** > >All: > >I am looking for a way to fluorescently counter-stain the volume of >cultured >cells, with a result similar to what would be achieved by having them >express GFP in the cytoplasm. I have seen some references to staining >with >eosin Y. Does anyone have other suggestions? > >Thanks in advance. > >Eric |
Free forum by Nabble | Edit this page |