Green fluorescing counterstain?

classic Classic list List threaded Threaded
♦
♦
7 messages Options
Stanislav Vitha Stanislav Vitha
Reply | Threaded
Open this post in threaded view
♦
♦
|

Green fluorescing counterstain?

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear confocalists,
do you have suggestions for a general counterstain that would fluoresce in
green?
I am doing 3D imaging of large number of pollen samples, and would like to
increase the signal levels so that my acquisition times become more
reasonable.

The pollen has been processed (cytoplasm removed), so I am trying to stain
the shell (exine/intine) and am not interested in the cytoplasm.

Rhodamine B staining worked fine, but the emission is in red. I am after
shorter wavelength signal for best resolution.

Thank you!


Stan Vitha
Deborah van der List Deborah van der List
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Green fluorescing counterstain?

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Stan,
Try  SYBR green II from Molecular Probes. I use it as a nuclear stain on
fixed tissue and use it just as I would with DAPI. I use the Argon laser to
scan.   I don't know how it will work on your pollen though.
Deborah

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Stanislav Vitha
Sent: Wednesday, January 30, 2008 4:02 PM
To: [hidden email]
Subject: Green fluorescing counterstain?

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear confocalists,
do you have suggestions for a general counterstain that would fluoresce in
green?
I am doing 3D imaging of large number of pollen samples, and would like to
increase the signal levels so that my acquisition times become more
reasonable.

The pollen has been processed (cytoplasm removed), so I am trying to stain
the shell (exine/intine) and am not interested in the cytoplasm.

Rhodamine B staining worked fine, but the emission is in red. I am after
shorter wavelength signal for best resolution.

Thank you!


Stan Vitha
Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Green fluorescing counterstain?

In reply to this post by Stanislav Vitha
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

You could try safranin.  Seems to stain lots of things and
excites well at 488 (even better at 514).  Fluorescence peaks
in the yellow but it's pretty broad.  I believe the old Bio-Rad
paper test slides were stained with safranin.

                                                       Guy

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Stanislav Vitha
Sent: Thursday, 31 January 2008 11:02 AM
To: [hidden email]
Subject: Green fluorescing counterstain?

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear confocalists,
do you have suggestions for a general counterstain that would fluoresce
in
green?
I am doing 3D imaging of large number of pollen samples, and would like
to
increase the signal levels so that my acquisition times become more
reasonable.

The pollen has been processed (cytoplasm removed), so I am trying to
stain
the shell (exine/intine) and am not interested in the cytoplasm.

Rhodamine B staining worked fine, but the emission is in red. I am after

shorter wavelength signal for best resolution.

Thank you!


Stan Vitha

No virus found in this incoming message.
Checked by AVG Free Edition.
Version: 7.5.516 / Virus Database: 269.19.17/1252 - Release Date:
30/01/2008 8:51 PM
 
George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Green fluorescing counterstain?

In reply to this post by Stanislav Vitha
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Reflected light confocal microscopy is much brighter than using a
fluorescent dye.




At 07:02 PM 1/30/2008, you wrote:

>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>Dear confocalists,
>do you have suggestions for a general counterstain that would fluoresce in
>green?
>I am doing 3D imaging of large number of pollen samples, and would like to
>increase the signal levels so that my acquisition times become more
>reasonable.
>
>The pollen has been processed (cytoplasm removed), so I am trying to stain
>the shell (exine/intine) and am not interested in the cytoplasm.
>
>Rhodamine B staining worked fine, but the emission is in red. I am after
>shorter wavelength signal for best resolution.
>
>Thank you!
>
>
>Stan Vitha






George McNamara, Ph.D.
University of Miami, Miller School of Medicine
Image Core
Miami, FL 33010
[hidden email]
[hidden email]
305-243-8436 office
http://home.earthlink.net/~pubspectra/
http://home.earthlink.net/~geomcnamara/
http://www.sylvester.org/research/SR_lab_analytical.asp?ana=desc (see
Analytical Imaging Core Facility)
Boswell, Carl A - (cboswell) Boswell, Carl A - (cboswell)
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Green fluorescing counterstain?

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Stan,
I've had the experience of general cell staining with plain old FITC in a
buffer.  Since the material was stained with unconjugated rhodamine B, you
might give unconjugated FITC a try.

Carl

Carl A. Boswell, Ph.D.
Molecular and Cellular Biology
University of Arizona
520-954-7053
FAX 520-621-3709
----- Original Message -----
From: "George McNamara" <[hidden email]>
To: <[hidden email]>
Sent: Wednesday, January 30, 2008 8:51 PM
Subject: Re: Green fluorescing counterstain?


> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Reflected light confocal microscopy is much brighter than using a
> fluorescent dye.
>
>
>
>
> At 07:02 PM 1/30/2008, you wrote:
>>Search the CONFOCAL archive at
>>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>>Dear confocalists,
>>do you have suggestions for a general counterstain that would fluoresce in
>>green?
>>I am doing 3D imaging of large number of pollen samples, and would like to
>>increase the signal levels so that my acquisition times become more
>>reasonable.
>>
>>The pollen has been processed (cytoplasm removed), so I am trying to stain
>>the shell (exine/intine) and am not interested in the cytoplasm.
>>
>>Rhodamine B staining worked fine, but the emission is in red. I am after
>>shorter wavelength signal for best resolution.
>>
>>Thank you!
>>
>>
>>Stan Vitha
>
>
>
>
>
>
> George McNamara, Ph.D.
> University of Miami, Miller School of Medicine
> Image Core
> Miami, FL 33010
> [hidden email]
> [hidden email]
> 305-243-8436 office
> http://home.earthlink.net/~pubspectra/
> http://home.earthlink.net/~geomcnamara/
> http://www.sylvester.org/research/SR_lab_analytical.asp?ana=desc (see
> Analytical Imaging Core Facility)
>
Stanislav Vitha Stanislav Vitha
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Green fluorescing counterstain?

In reply to this post by Stanislav Vitha
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

not if you look at an almost vertical feature, e.g., the side of a sphere
(or a pollen grain)

Stan

On Wed, 30 Jan 2008 22:51:40 -0500, George McNamara
<[hidden email]> wrote:

>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>Reflected light confocal microscopy is much brighter than using a
>fluorescent dye.
>
>
>
>
>At 07:02 PM 1/30/2008, you wrote:
>>Search the CONFOCAL archive at
>>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>>Dear confocalists,
>>do you have suggestions for a general counterstain that would fluoresce
in
>>green?
>>I am doing 3D imaging of large number of pollen samples, and would like
to
>>increase the signal levels so that my acquisition times become more
>>reasonable.
>>
>>The pollen has been processed (cytoplasm removed), so I am trying to
stain

>>the shell (exine/intine) and am not interested in the cytoplasm.
>>
>>Rhodamine B staining worked fine, but the emission is in red. I am after
>>shorter wavelength signal for best resolution.
>>
>>Thank you!
>>
>>
>>Stan Vitha
>
>
>
>
>
>
>George McNamara, Ph.D.
>University of Miami, Miller School of Medicine
>Image Core
>Miami, FL 33010
>[hidden email]
>[hidden email]
>305-243-8436 office
>http://home.earthlink.net/~pubspectra/
>http://home.earthlink.net/~geomcnamara/
>http://www.sylvester.org/research/SR_lab_analytical.asp?ana=desc (see
>Analytical Imaging Core Facility)
>=========================================================================
Judy Trogadis Judy Trogadis
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Green fluorescing counterstain?

In reply to this post by Guy Cox
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

This is a late reply but I just saw a brief article in the latest Microscopy Today (Vol 16(1):12-15) where they describe whole cell fluorescent stains, without any specificity, and available with 4 different Ex/Em spectra. Sounds like a counterstain to me.

Made by Cellomics, a subsidiary of Thermo Fisher Scientific.  I think I will try one of the flavours myself to see how well it works.

Judy



Judy Trogadis
Bio-Imaging Coordinator
St. Michael's Hospital, 7Queen
30 Bond St.
Toronto, ON M5B 1W8, Canada
ph:  416-864-6060  x6337
pager: 416-685-9219
fax: 416-864-6043
[hidden email]


>>> Guy Cox <[hidden email]> 01/30/08 7:53 PM >>>
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal 

You could try safranin.  Seems to stain lots of things and
excites well at 488 (even better at 514).  Fluorescence peaks
in the yellow but it's pretty broad.  I believe the old Bio-Rad
paper test slides were stained with safranin.

                                                       Guy

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Stanislav Vitha
Sent: Thursday, 31 January 2008 11:02 AM
To: [hidden email]
Subject: Green fluorescing counterstain?

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal 

Dear confocalists,
do you have suggestions for a general counterstain that would fluoresce
in
green?
I am doing 3D imaging of large number of pollen samples, and would like
to
increase the signal levels so that my acquisition times become more
reasonable.

The pollen has been processed (cytoplasm removed), so I am trying to
stain
the shell (exine/intine) and am not interested in the cytoplasm.

Rhodamine B staining worked fine, but the emission is in red. I am after

shorter wavelength signal for best resolution.

Thank you!


Stan Vitha