HBO/Mercury lamp lifetimes

classic Classic list List threaded Threaded
52 messages Options
123
Anda Cornea Anda Cornea
Reply | Threaded
Open this post in threaded view
|

Re: HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

So you have time for this stuff, but not for me?  Gotcha!

Anda Cornea Ph.D.
Head of the Imaging and Morphology Core
Oregon National Primate Research Center
Oregon Health & Science University
(503) 690-5293
>>> [hidden email] 04/25/08 1:39 PM >>>
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

We recently encountered a failure of a Nikon Xenon housing on a  
vintage 1992 Diaphot.  The hard plastic insulation on part of the  
cable interface broke down and arced to the lamp housing at  
ignition.  The users reported that it kept clicking (sparking),  
wouldn't ignite and the burning odor wouldn't go away.  Gee, at least  
they finally got the idea and turned it off.

Our only other non-bulb issue was a Prior power supply on a Leica  
MZ-12 FLIII that Prior never could get to keep igniting.

  I change our Hg bulbs about 300-320 hrs  on widefield systems, but  
let a confocal go to 350-380 since it has fewer ignitions.  The other  
confocal has a 75W xenon with about 4,000 hours.

Ushio 102D bulbs have been much more reliable than the Osram 103W,  
especially with older power supplies.    I've replaced all of the  
Osram bulbs with Ushio.

Glen

Glen MacDonald
Core for Communication Research
Virginia Merrill Bloedel Hearing Research Center
Box 357923
University of Washington
Seattle, WA 98195-7923  USA
(206) 616-4156
[hidden email]

************************************************************************

******
The box said "Requires WindowsXP or better", so I bought a Macintosh.
************************************************************************

******


On Apr 25, 2008, at 10:41 AM, Ignatius, Mike wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> An important side bar in this discussion, is what warning signs are
> there that the bulb is too old and possibly going to fail/explode.
>
> For me, the earliest signs that the arc gap has grown to impossible
> widths were difficulty in starting the bulb to begin with, flickering
> when on, long time to stabilize etc.
>
> The more dreadful indication is an electrical burning smell when it is
> on and about to blow.
>
> In my one experience it smelled like insulation burning off of wiring.
> Apparently the bulb super heats before failure?  I would be curious to
> hear if others have had this warning sign before the bulb blew.  I
> stupidly got up, and sniffed around the scope to see what the  
> source was
> and wham the bulb blew in my face.  I thanked Nikon for their sturdy
> housing, no glass came through, though the diffuser lens inside did
> shatter, and the mercury vapor surely escaped.  After it has blown,
> always good to leave the area for a while to let the gas escape.
>
> My recommendation: if you smell electrical burning, turn off the lamp
> and check your bulb!  If the smell persists, something else is amiss.
>
> And no one has mentioned benefits to running cooling fans across  
> housing
> this time around.  It seems to help extend lamp life.
>
> Mike Ignatius
>
> -----Original Message-----
> From: Confocal Microscopy List  
> [mailto:[hidden email]] On
> Behalf Of Russell Spear
> Sent: Friday, April 25, 2008 7:58 AM
> To: [hidden email]
> Subject: Re: HBO/Mercury lamp lifetimes
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> I've been using HBO 100 W s, for about 30 years, and have had 2
> explosions in that time, one long ago and one approx. 3 yrs past.  
> Both
> could be attributed to a high number of on off cycles (something I
> discourage).  I nomally cycle a bulb no more than twice a day, and  
> need
> to relamp about every 40-45 days in a busy lab with 3 fluorescent
> scopes.  We always replace at 200 hrs. the cost of the bulb is small
> compared to the damage to the equipment and time and cost of
> decontaminating the microscope rooms.
>
> Russ
>
>
>
> Russell N. Spear
> Sr. Research Specialist
> Dept. of Plant Pathology
> Univ. of Wisconsin
> 1630 Linden Dr.
> Madison WI 53706
>
> voice 608.263.2093
> fax     608.263.2626
Stephen Cody Stephen Cody
Reply | Threaded
Open this post in threaded view
|

Re: HBO/Mercury lamp lifetimes

In reply to this post by George Ring
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear George,

 

Photos of the damage would be a nice teaching aid. Did you happen to take any photos?

 

Cheers

Stephen H. Cody
Microscopy Manager
Central Resource for Advanced Microscopy
Ludwig Institute for Cancer Research

PO Box 2008 Royal Melbourne Hospital
Victoria,      3050
Australia
Tel: 61 3 9341 3155    Fax: 61 3 9341 3104
email: [hidden email]
www.ludwig.edu.au/labs/confocal.html
www.ludwig.edu.au/confocal

Tip: Learn how to receive reminders about you microscope booking:
www.ludwig.edu.au/confocal/Local/Booking_Hint.htm
 

-----Original Message-----
From: Confocal Microscopy List [mailto:
[hidden email]] On Behalf Of George Ring
Sent: Friday, 25 April 2008 11:42 PM
To:
[hidden email]
Subject: Re: HBO/Mercury lamp lifetimes

 

We change our bulb between 300 and 400 hours.  We had a mercury bulb explode once several years ago,  and it caused at least a couple  thousand dollars damage to the microscope (condenser lens, housing, etc).  To me, the savings of $50/yr by letting the bulb go longer isn't worth the potential cost in money and downtime of an explosion.

...just my 2 cents worth.

 

George

 

George Ring, Ph.D.
Dept. of Cell and Developmental Biology
SUNY Upstate Medical University
750 E. Adams St.
Syracuse NY  13210
Tel. (315) 464-8595
FAX (315) 464-8535
email: [hidden email]

>>> Graham Wright <[hidden email]> 4/24/08 9:30:11 PM >>>

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hello,

This is not strictly a confocal question, but reasonably relevant and it's not been recently discussed from what I can tell from the archives.

I was hoping to find out at how many hours people change the HBO/Mercury short arc lamps (specifically Osram 103W). The microscope manufacturers (Zeiss & Leica) always recommend 300 hours, due to the potential dangers of explosion when using them for longer... but what do users/facilities tend to do?

Someone recently told me that the lamps are now made in such a way that explosions are less common that ~10 years ago (when discussion on this list regarding explosions were more common)

Thanks in advance,

Graham

 

-----
Dr Graham Wright
Microscopy & Imaging Facility

 

Temasek Life Sciences Laboratory
1 Research Link
National University of Singapore
Singapore 117604

 

P:    +65 6872 8406
M:   +65 8256 7916
E:     [hidden email]
W:    www.bioinformatics.tll.org.sg/labs/microscopy/index.htm


This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research does not waiver any rights if you have received this communication in error.
The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research.

George Ring George Ring
Reply | Threaded
Open this post in threaded view
|

Re: HBO/Mercury lamp lifetimes

Dear Steve,
 
No, sorry, no pictures.  Just a big bill.
 
George
 
 
George Ring, Ph.D.
Dept. of Cell and Developmental Biology
SUNY Upstate Medical University
750 E. Adams St.
Syracuse NY  13210
Tel. (315) 464-8595
FAX (315) 464-8535
email: [hidden email]

>>> Stephen Cody <[hidden email]> 4/27/08 10:47:35 PM >>>
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear George,

 

Photos of the damage would be a nice teaching aid. Did you happen to take any photos?

 

Cheers

Stephen H. Cody
Microscopy Manager
Central Resource for Advanced Microscopy
Ludwig Institute for Cancer Research

PO Box 2008 Royal Melbourne Hospital
Victoria,      3050
Australia
Tel: 61 3 9341 3155    Fax: 61 3 9341 3104
email: [hidden email]
www.ludwig.edu.au/labs/confocal.html
www.ludwig.edu.au/confocal

Tip: Learn how to receive reminders about you microscope booking:
www.ludwig.edu.au/confocal/Local/Booking_Hint.htm
 

-----Original Message-----
From: Confocal Microscopy List [mailto:
[hidden email]] On Behalf Of George Ring
Sent: Friday, 25 April 2008 11:42 PM
To:
[hidden email]
Subject: Re: HBO/Mercury lamp lifetimes

 

We change our bulb between 300 and 400 hours.  We had a mercury bulb explode once several years ago,  and it caused at least a couple  thousand dollars damage to the microscope (condenser lens, housing, etc).  To me, the savings of $50/yr by letting the bulb go longer isn't worth the potential cost in money and downtime of an explosion.

...just my 2 cents worth.

 

George

 

George Ring, Ph.D.
Dept. of Cell and Developmental Biology
SUNY Upstate Medical University
750 E. Adams St.
Syracuse NY  13210
Tel. (315) 464-8595
FAX (315) 464-8535
email: [hidden email]

>>> Graham Wright <[hidden email]> 4/24/08 9:30:11 PM >>>

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hello,

This is not strictly a confocal question, but reasonably relevant and it's not been recently discussed from what I can tell from the archives.

I was hoping to find out at how many hours people change the HBO/Mercury short arc lamps (specifically Osram 103W). The microscope manufacturers (Zeiss & Leica) always recommend 300 hours, due to the potential dangers of explosion when using them for longer... but what do users/facilities tend to do?

Someone recently told me that the lamps are now made in such a way that explosions are less common that ~10 years ago (when discussion on this list regarding explosions were more common)

Thanks in advance,

Graham

 

-----
Dr Graham Wright
Microscopy & Imaging Facility

 

Temasek Life Sciences Laboratory
1 Research Link
National University of Singapore
Singapore 117604

 

P:    +65 6872 8406
M:   +65 8256 7916
E:     [hidden email]
W:    www.bioinformatics.tll.org.sg/labs/microscopy/index.htm


This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research does not waiver any rights if you have received this communication in error.
The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research.

Karen Lovely-Leach Karen Lovely-Leach
Reply | Threaded
Open this post in threaded view
|

Re: HBO/Mercury lamp lifetimes

In reply to this post by Dale Callaham
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Karen Lovely-Leach
MacGuire LSB- Room 326
Amherst College
Amherst, MA  01002
413-542-2081
[hidden email]

Hi everyone,
I started running our new confocal facility a little over a year ago
with an LSM5 Pascal. Last year we got the recommended 200 hours out of
the HBO100 W/2, after which it began to stutter on ignition, and we
changed it promptly.  Because we use it most of the time to locate cells
and quickly look for staining, we started running it at 20%, and since
then we have logged approximately 750 hours.  Now I am getting nervous
with thoughts of explosions, so I am going to replace it!
I found all of the conversation on this topic very helpful.  Karen
 
-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Dale Callaham
Sent: Friday, April 25, 2008 6:07 PM
To: [hidden email]
Subject: Re: HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi,

I have seen strong recommendations against forced air cooling where arc
stability is a prime concern. Osram literature states that forced air is

not needed if sufficient convection through the lamp housing will keep
the bases below 230C and most commercial designs probably do this. They
state that any cooling should "not be directed at the discharge vessel".

Most recent lamp housings and lamp mountings assist the upper base
cooling with a heatsink. I bolted a thermocouple to the upper base (-
terminal) of a unit (leave TC disconnected until after ignition!) and
measured only 180C for a new lamp; the lamp has to get hot to function
properly. Staying with well-designed equipment in good operating
condition (check for oxidation of leads and connectors and replace if
necessary) it shouldn't be necessary to use forced air.

I am curious - someone mentioned the Zeiss AttoArc - doesn't that allow
you to run at lower than max/nominal settings? Seems that many arclamps
are too bright and need attenuation anyway; I was wondering if anyone
runs them at lower levels as a standby condition between users or for
lower excitation and does it seem to extend the life? I built a unit
that could run a HBO 100W/2 nicely below 1A (they have a 5A nominal
current) but I didn't get to monitor the performance over a long period.

Dale

Ignatius, Mike wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> An important side bar in this discussion, is what warning signs are
> there that the bulb is too old and possibly going to fail/explode.  
>
> For me, the earliest signs that the arc gap has grown to impossible
> widths were difficulty in starting the bulb to begin with, flickering
> when on, long time to stabilize etc.  
>
> The more dreadful indication is an electrical burning smell when it is
> on and about to blow.
>
> In my one experience it smelled like insulation burning off of wiring.
> Apparently the bulb super heats before failure?  I would be curious to
> hear if others have had this warning sign before the bulb blew.  I
> stupidly got up, and sniffed around the scope to see what the source
was
> and wham the bulb blew in my face.  I thanked Nikon for their sturdy
> housing, no glass came through, though the diffuser lens inside did
> shatter, and the mercury vapor surely escaped.  After it has blown,
> always good to leave the area for a while to let the gas escape.  
>
> My recommendation: if you smell electrical burning, turn off the lamp
> and check your bulb!  If the smell persists, something else is amiss.

>
> And no one has mentioned benefits to running cooling fans across
housing
> this time around.  It seems to help extend lamp life.
>
> Mike Ignatius
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]]
On

> Behalf Of Russell Spear
> Sent: Friday, April 25, 2008 7:58 AM
> To: [hidden email]
> Subject: Re: HBO/Mercury lamp lifetimes
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> I've been using HBO 100 W s, for about 30 years, and have had 2
> explosions in that time, one long ago and one approx. 3 yrs past.
Both
> could be attributed to a high number of on off cycles (something I
> discourage).  I nomally cycle a bulb no more than twice a day, and
need

> to relamp about every 40-45 days in a busy lab with 3 fluorescent
> scopes.  We always replace at 200 hrs. the cost of the bulb is small
> compared to the damage to the equipment and time and cost of
> decontaminating the microscope rooms.
>
> Russ
>
>
>
> Russell N. Spear
> Sr. Research Specialist
> Dept. of Plant Pathology
> Univ. of Wisconsin
> 1630 Linden Dr.
> Madison WI 53706
>
> voice 608.263.2093
> fax     608.263.2626
Ian Dobbie-2 Ian Dobbie-2
Reply | Threaded
Open this post in threaded view
|

Re: HBO/Mercury lamp lifetimes

In reply to this post by Nowell, Cameron
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

"Nowell, Cameron" <[hidden email]> writes:

> We have been doing this for a year and a half now and have not had
> any problems. When we first started to do it, we asked a lot of
> people about the exploding lamps. No one we spoke to had had it
> happen to them but everyone new someone (or new someone who new
> someone) who had had it happen in the past.

At one of my previous jobs we had a 50W Zeiss arc lamp explode 9sorry
I don't know the model number). This was on a rarely used scope used
to check staining or tansfection efficiencies, and so lamp stability or
intensity was not important. I expect the lamp had done multiple
times its recommended lifetime. The result was pretty dramatic, the
reflector was shattered and the collection lens damaged. We ditched
the lamp housing after this, not worth fixing (or decontaminating).

Anybody know how dangerous the mercury vapour from this type of
explosion would be? We evacuated the room for a few hours, didn't seem
to be much else we could do.


Ian
Russell Spear Russell Spear
Reply | Threaded
Open this post in threaded view
|

Re: HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Ian

Last time we had a mishap like this our safety dept. banned all use of
the room until they decontaminated all surfaces with an absorbent and
monitored for Hg vapor for 72hrs.  Then gave us stern lecturing and
billed us.  Room was off limits for about 1 1/2 wks.

So it isn't worth the risk to use them past rated hours.

Russ



Russell N. Spear
Sr. Research Specialist
Dept. of Plant Pathology
Univ. of Wisconsin
1630 Linden Dr.
Madison WI 53706

voice 608.263.2093
fax     608.263.2626

>>> Ian Dobbie <[hidden email]> 5/3/2008 9:53 AM >>>
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal 

"Nowell, Cameron" <[hidden email]> writes:

> We have been doing this for a year and a half now and have not had
> any problems. When we first started to do it, we asked a lot of
> people about the exploding lamps. No one we spoke to had had it
> happen to them but everyone new someone (or new someone who new
> someone) who had had it happen in the past.

At one of my previous jobs we had a 50W Zeiss arc lamp explode 9sorry
I don't know the model number). This was on a rarely used scope used
to check staining or tansfection efficiencies, and so lamp stability
or
intensity was not important. I expect the lamp had done multiple
times its recommended lifetime. The result was pretty dramatic, the
reflector was shattered and the collection lens damaged. We ditched
the lamp housing after this, not worth fixing (or decontaminating).

Anybody know how dangerous the mercury vapour from this type of
explosion would be? We evacuated the room for a few hours, didn't seem
to be much else we could do.


Ian
Todorov, Ivan Todorov, Ivan
Reply | Threaded
Open this post in threaded view
|

Re: HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Re: HBO/Mercury lamp lifetimes
Hi everybody,
all these scary stories about blown up lamps sound fancy, but please have in mind that these things might happen with lamps very far away from expiration date. e.g. the only accident like that we had in our institution was with a lamp just 60 hours old ... We are playing fluorescence quite a bit, and the japanese lamps we are using right now last 500+ hours ...
All the best + a great weekend!
Yours
Ivan
 
Ivan Todorov 
City of Hope
Duarte, Ca


From: Confocal Microscopy List on behalf of Russell Spear
Sent: Сб 5/3/2008 2:54
To: [hidden email]
Subject: Re: HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Ian

Last time we had a mishap like this our safety dept. banned all use of
the room until they decontaminated all surfaces with an absorbent and
monitored for Hg vapor for 72hrs.  Then gave us stern lecturing and
billed us.  Room was off limits for about 1 1/2 wks.

So it isn't worth the risk to use them past rated hours.

Russ



Russell N. Spear
Sr. Research Specialist
Dept. of Plant Pathology
Univ. of Wisconsin
1630 Linden Dr.
Madison WI 53706

voice 608.263.2093
fax     608.263.2626

>>> Ian Dobbie <[hidden email]> 5/3/2008 9:53 AM >>>
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

"Nowell, Cameron" <[hidden email]> writes:

> We have been doing this for a year and a half now and have not had
> any problems. When we first started to do it, we asked a lot of
> people about the exploding lamps. No one we spoke to had had it
> happen to them but everyone new someone (or new someone who new
> someone) who had had it happen in the past.

At one of my previous jobs we had a 50W Zeiss arc lamp explode 9sorry
I don't know the model number). This was on a rarely used scope used
to check staining or tansfection efficiencies, and so lamp stability
or
intensity was not important. I expect the lamp had done multiple
times its recommended lifetime. The result was pretty dramatic, the
reflector was shattered and the collection lens damaged. We ditched
the lamp housing after this, not worth fixing (or decontaminating).

Anybody know how dangerous the mercury vapour from this type of
explosion would be? We evacuated the room for a few hours, didn't seem
to be much else we could do.


Ian


mg_info.txt (1K) Download Attachment
S. Pagakis (IIBEAA) S. Pagakis (IIBEAA)
Reply | Threaded
Open this post in threaded view
|

Re: HBO/Mercury lamp lifetimes

In reply to this post by Russell Spear
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hello everyone

I think we are barking under the wrong tree.  Please correct me if i  
am wrong, but many explosions I read reported in this thread (and  
only one -in 25 years- which happened to me @50hrs) happened at way  
less than the rated hours.

Shouldn't  we revise the myth that changing the lamp early protects  
from explosions?

It seems to me that it is either due to material failure (in which  
case I think the lamp manufacturers should pay the damage bill) or  
too many ignitions, in which case the manufacturers should include  
this count in the specs.

Regards,

*********************************
Stamatis Pagakis Ph.D.
Biological Imaging Unit
Biomedical Research Foundation, Academy of Athens,  Greece


> On 4 May 2008, at 00:54, Russell Spear wrote:
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> So it isn't worth the risk to use them past rated hours.
>
> Russ
>
>
>
> Russell N. Spear
> Sr. Research Specialist
> Dept. of Plant Pathology
> Univ. of Wisconsin
> 1630 Linden Dr.
> Madison WI 53706
>
> voice 608.263.2093
> fax     608.263.2626
>
James Pawley James Pawley
Reply | Threaded
Open this post in threaded view
|

Re: HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi all,

I would like to go back to an earlier post, and reiterate that there
are several different types of power supplies. The constant-current
ones increase the applied voltage as the "conductivity" of the bulb
decreases. This means that the amount of power deposited in the arc
slowly INCREASES as the electrodes age and eventually, this will
cause overheating, and softening of the fused-quartz envelope.
Constant-power supplies will not cause this problem, but the
intensity of the brightest part of the arc will slowly deteriorate
with time.

Another factor is dirt. Either finger grease or perhaps some
accidental event such as a small, fried insect or a very small piece
of flaked-off black paint. In short, anything that absorbs light and
gets hot while attached in some way to the quartz envelope. Any "dark
spot" will cause local overheating and asymmetrical envelope failure.

A second possibility has to to with the connections between the bulb
and the housing. If the mating surfaces of these metal parts become
corroded, or oxidized, the contact resistance will increase as will
the working temperature. This also can lead to premature envelope
failure.

Even the alignment of the spherical rear mirror can be important. If
the reflected image of the arc is formed not to the side of the arc
itself but concentrated onto a small section of the envelope, this
can exacerbate any asymmetrical heating cause by dirt or
contamination.

The effect of these "operator" factors will vary somewhat depending
on the type of power supply used.

Therefore, it seems to me that the list of important variables should
include not just be the bulb manufacturer, the bulb type  and the
design of the power supply but the care with which the new bulb has
been installed and aligned.

And then there is the whole matter of variable-power Hg systems. We
usually use our Atto-arc 50W at about 20% power and the bulbs often
seem to last for over a thousand hours. More than enough brightness
to find  the part of your sample best suited to confocal analysis.

Finally, there is the matter of xenon arcs. These have all the same
problems mentioned above, but in addition, because they are reputed
to start off having a much higher gas pressure, any resulting
explosions are said to be much more impressive, in fact even
dangerous (albeit, without the additional nuisance of Hg
contamination). Has anyone had any experience here?

Cheers,

Jim Pawley


>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>Hello everyone
>
>I think we are barking under the wrong tree.  Please correct me if i
>am wrong, but many explosions I read reported in this thread (and
>only one -in 25 years- which happened to me @50hrs) happened at way
>less than the rated hours.
>
>Shouldn't  we revise the myth that changing the lamp early protects
>from explosions?
>
>It seems to me that it is either due to material failure (in which
>case I think the lamp manufacturers should pay the damage bill) or
>too many ignitions, in which case the manufacturers should include
>this count in the specs.
>
>Regards,
>
>*********************************
>Stamatis Pagakis Ph.D.
>Biological Imaging Unit
>Biomedical Research Foundation, Academy of Athens,  Greece
>
>>On 4 May 2008, at 00:54, Russell Spear wrote:
>>Search the CONFOCAL archive at
>>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>>So it isn't worth the risk to use them past rated hours.
>>
>>Russ
>>
>>
>>
>>Russell N. Spear
>>Sr. Research Specialist
>>Dept. of Plant Pathology
>>Univ. of Wisconsin
>>1630 Linden Dr.
>>Madison WI 53706
>>
>>voice 608.263.2093
>>fax     608.263.2626


--
               **********************************************
Prof. James B. Pawley,                          Ph.  608-263-3147
Room 223, Zoology Research Building,              
FAX  608-265-5315
1117 Johnson Ave., Madison, WI, 53706  
[hidden email]
3D Microscopy of Living Cells Course, June 14-26, 2008, UBC, Vancouver Canada
Info: http://www.3dcourse.ubc.ca/             Applications due by March 15, 2008
               "If it ain't diffraction, it must be statistics." Anon.
Jim Beacher Jim Beacher
Reply | Threaded
Open this post in threaded view
|

Re: HBO/Mercury lamp lifetimes

In reply to this post by Graham Wright-5
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Re: HBO/Mercury lamp lifetimes
Commercial Interest
 
 
Having followed this discussion, we can contain ourselves no longer.........LEDs!
 
 
 
JIM Beacher
CoolLED Ltd
 
 
-----Original Message-----
From: "Confocal Microscopy List" <[hidden email]>
Sent: 07/05/2008 14:50
To: "[hidden email]" <[hidden email]>
Subject: Re: HBO/Mercury lamp lifetimes


Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi all,

I would like to go back to an earlier post, and reiterate that there
are several different types of power supplies. The constant-current
ones increase the applied voltage as the "conductivity" of the bulb
decreases. This means that the amount of power deposited in the arc
slowly INCREASES as the electrodes age and eventually, this will
cause overheating, and softening of the fused-quartz envelope.
Constant-power supplies will not cause this problem, but the
intensity of the brightest part of the arc will slowly deteriorate
with time.

Another factor is dirt. Either finger grease or perhaps some
accidental event such as a small, fried insect or a very small piece
of flaked-off black paint. In short, anything that absorbs light and
gets hot while attached in some way to the quartz envelope. Any "dark
spot" will cause local overheating and asymmetrical envelope failure.

A second possibility has to to with the connections between the bulb
and the housing. If the mating surfaces of these metal parts become
corroded, or oxidized, the contact resistance will increase as will
the working temperature. This also can lead to premature envelope
failure.

Even the alignment of the spherical rear mirror can be important. If
the reflected image of the arc is formed not to the side of the arc
itself but concentrated onto a small section of the envelope, this
can exacerbate any asymmetrical heating cause by dirt or
contamination.

The effect of these "operator" factors will vary somewhat depending
on the type of power supply used.

Therefore, it seems to me that the list of important variables should
include not just be the bulb manufacturer, the bulb type  and the
design of the power supply but the care with which the new bulb has
been installed and aligned.

And then there is the whole matter of variable-power Hg systems. We
usually use our Atto-arc 50W at about 20% power and the bulbs often
seem to last for over a thousand hours. More than enough brightness
to find  the part of your sample best suited to confocal analysis.

Finally, there is the matter of xenon arcs. These have all the same
problems mentioned above, but in addition, because they are reputed
to start off having a much higher gas pressure, any resulting
explosions are said to be much more impressive, in fact even
dangerous (albeit, without the additional nuisance of Hg
contamination). Has anyone had any experience here?

Cheers,

Jim Pawley


>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>Hello everyone
>
>I think we are barking under the wrong tree.  Please correct me if i
>am wrong, but many explosions I read reported in this thread (and
>only one -in 25 years- which happened to me @50hrs) happened at way
>less than the rated hours.
>
>Shouldn't  we revise the myth that changing the lamp early protects
>from explosions?
>
>It seems to me that it is either due to material failure (in which
>case I think the lamp manufacturers should pay the damage bill) or
>too many ignitions, in which case the manufacturers should include
>this count in the specs.
>
>Regards,
>
>*********************************
>Stamatis Pagakis Ph.D.
>Biological Imaging Unit
>Biomedical Research Foundation, Academy of Athens,  Greece
>
>>On 4 May 2008, at 00:54, Russell Spear wrote:
>>Search the CONFOCAL archive at
>>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>>So it isn't worth the risk to use them past rated hours.
>>
>>Russ
>>
>>
>>
>>Russell N. Spear
>>Sr. Research Specialist
>>Dept. of Plant Pathology
>>Univ. of Wisconsin
>>1630 Linden Dr.
>>Madison WI 53706
>>
>>voice 608.263.2093
>>fax     608.263.2626


--
               **********************************************
Prof. James B. Pawley,                                      Ph.  608-263-3147
Room 223, Zoology Research Building,              
FAX  608-265-5315
1117 Johnson Ave., Madison, WI, 53706 
[hidden email]
3D Microscopy of Living Cells Course, June 14-26, 2008, UBC, Vancouver Canada
Info: http://www.3dcourse.ubc.ca/            Applications due by March 15, 2008
               "If it ain't diffraction, it must be statistics." Anon.

Alison J. North Alison J. North
Reply | Threaded
Open this post in threaded view
|

Re: HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Jim,

Oh you cruel man, you are mercilessly TAUNTING those of us based in the
USA who would love to buy your LEDs but are prevented from doing so by
exceptionally irritating patent restrictions (hello to the company
concerned - your LED system is prohibitively expensive!). Or am I
missing something and this situation has recently changed?  Do keep us
posted please?  And yes, I am intentionally sending this to the
listserver, so that other people based in this country don't get all
excited by your LEDs and have their hopes dashed, like us.  Plus, if I
am incorrect about the patent situation and we can in fact obtain an
LED-based lamp here at a reasonable price (without having to assemble
everything ourselves), we would like to know.

Thanks and best wishes,
Alison


James Beacher wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
> Commercial Interest
>  
>  
> Having followed this discussion, we can contain ourselves no
> longer.........*LEDs!*
>  
>  
>  
> JIM Beacher
> CoolLED Ltd
>  
>  
>
>     -----Original Message-----
>     *From:* "Confocal Microscopy List" <[hidden email]>
>     *Sent:* 07/05/2008 14:50
>     *To:* "[hidden email]" <[hidden email]>
>     *Subject:* Re: HBO/Mercury lamp lifetimes
>
>
>     Search the CONFOCAL archive at
>     http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>     Hi all,
>
>     I would like to go back to an earlier post, and reiterate that there
>     are several different types of power supplies. The constant-current
>     ones increase the applied voltage as the "conductivity" of the bulb
>     decreases. This means that the amount of power deposited in the arc
>     slowly INCREASES as the electrodes age and eventually, this will
>     cause overheating, and softening of the fused-quartz envelope.
>     Constant-power supplies will not cause this problem, but the
>     intensity of the brightest part of the arc will slowly deteriorate
>     with time.
>
>     Another factor is dirt. Either finger grease or perhaps some
>     accidental event such as a small, fried insect or a very small piece
>     of flaked-off black paint. In short, anything that absorbs light and
>     gets hot while attached in some way to the quartz envelope. Any "dark
>     spot" will cause local overheating and asymmetrical envelope failure.
>
>     A second possibility has to to with the connections between the bulb
>     and the housing. If the mating surfaces of these metal parts become
>     corroded, or oxidized, the contact resistance will increase as will
>     the working temperature. This also can lead to premature envelope
>     failure.
>
>     Even the alignment of the spherical rear mirror can be important. If
>     the reflected image of the arc is formed not to the side of the arc
>     itself but concentrated onto a small section of the envelope, this
>     can exacerbate any asymmetrical heating cause by dirt or
>     contamination.
>
>     The effect of these "operator" factors will vary somewhat depending
>     on the type of power supply used.
>
>     Therefore, it seems to me that the list of important variables should
>     include not just be the bulb manufacturer, the bulb type  and the
>     design of the power supply but the care with which the new bulb has
>     been installed and aligned.
>
>     And then there is the whole matter of variable-power Hg systems. We
>     usually use our Atto-arc 50W at about 20% power and the bulbs often
>     seem to last for over a thousand hours. More than enough brightness
>     to find  the part of your sample best suited to confocal analysis.
>
>     Finally, there is the matter of xenon arcs. These have all the same
>     problems mentioned above, but in addition, because they are reputed
>     to start off having a much higher gas pressure, any resulting
>     explosions are said to be much more impressive, in fact even
>     dangerous (albeit, without the additional nuisance of Hg
>     contamination). Has anyone had any experience here?
>
>     Cheers,
>
>     Jim Pawley
>
>
>      >Search the CONFOCAL archive at
>      >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>      >
>      >Hello everyone
>      >
>      >I think we are barking under the wrong tree.  Please correct me if i
>      >am wrong, but many explosions I read reported in this thread (and
>      >only one -in 25 years- which happened to me @50hrs) happened at way
>      >less than the rated hours.
>      >
>      >Shouldn't  we revise the myth that changing the lamp early protects
>      >from explosions?
>      >
>      >It seems to me that it is either due to material failure (in which
>      >case I think the lamp manufacturers should pay the damage bill) or
>      >too many ignitions, in which case the manufacturers should include
>      >this count in the specs.
>      >
>      >Regards,
>      >
>      >*********************************
>      >Stamatis Pagakis Ph.D.
>      >Biological Imaging Unit
>      >Biomedical Research Foundation, Academy of Athens,  Greece
>      >
>      >>On 4 May 2008, at 00:54, Russell Spear wrote:
>      >>Search the CONFOCAL archive at
>      >>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>      >>
>      >>So it isn't worth the risk to use them past rated hours.
>      >>
>      >>Russ
>      >>
>      >>
>      >>
>      >>Russell N. Spear
>      >>Sr. Research Specialist
>      >>Dept. of Plant Pathology
>      >>Univ. of Wisconsin
>      >>1630 Linden Dr.
>      >>Madison WI 53706
>      >>
>      >>voice 608.263.2093
>      >>fax     608.263.2626
>
>
>     --
>                    **********************************************
>     Prof. James B. Pawley,                                      Ph.
>     608-263-3147
>     Room 223, Zoology Research Building,              
>     FAX  608-265-5315
>     1117 Johnson Ave., Madison, WI, 53706
>     [hidden email]
>     3D Microscopy of Living Cells Course, June 14-26, 2008, UBC,
>     Vancouver Canada
>     Info: http://www.3dcourse.ubc.ca/            Applications due by
>     March 15, 2008
>                    "If it ain't diffraction, it must be statistics." Anon.
>

--
Alison J. North, Ph.D.,
Research Assistant Professor and
Director of the Bio-Imaging Resource Center,
The Rockefeller University,
1230 York Avenue,
New York,
NY 10065.
Tel: office ++ 212 327 7488
Tel: lab     ++ 212 327 7486
Fax:         ++ 212 327 7489
Tim Murphy-5 Tim Murphy-5
Reply | Threaded
Open this post in threaded view
|

LED source

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

for cheap mounted LEDs/drivers for microscopes you can try Thorlabs

http://www.thorlabs.com/NewGroupPage9.cfm?ObjectGroup_ID=2692&visNavID=551

I have no commercial interest here, they dont seem to have a 3-color system
yet but its a matter of time, maybe request it?

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]]On
Behalf Of Alison North
Sent: Wednesday, May 07, 2008 7:51 AM
To: [hidden email]
Subject: Re: HBO/Mercury lamp lifetimes


Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Jim,

Oh you cruel man, you are mercilessly TAUNTING those of us based in the
USA who would love to buy your LEDs but are prevented from doing so by
exceptionally irritating patent restrictions (hello to the company
concerned - your LED system is prohibitively expensive!). Or am I
missing something and this situation has recently changed?  Do keep us
posted please?  And yes, I am intentionally sending this to the
listserver, so that other people based in this country don't get all
excited by your LEDs and have their hopes dashed, like us.  Plus, if I
am incorrect about the patent situation and we can in fact obtain an
LED-based lamp here at a reasonable price (without having to assemble
everything ourselves), we would like to know.

Thanks and best wishes,
Alison


James Beacher wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
> Commercial Interest
>
>
> Having followed this discussion, we can contain ourselves no
> longer.........*LEDs!*
>
>
>
> JIM Beacher
> CoolLED Ltd
>
>
>
>     -----Original Message-----
>     *From:* "Confocal Microscopy List" <[hidden email]>
>     *Sent:* 07/05/2008 14:50
>     *To:* "[hidden email]" <[hidden email]>
>     *Subject:* Re: HBO/Mercury lamp lifetimes
>
>
>     Search the CONFOCAL archive at
>     http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>     Hi all,
>
>     I would like to go back to an earlier post, and reiterate that there
>     are several different types of power supplies. The constant-current
>     ones increase the applied voltage as the "conductivity" of the bulb
>     decreases. This means that the amount of power deposited in the arc
>     slowly INCREASES as the electrodes age and eventually, this will
>     cause overheating, and softening of the fused-quartz envelope.
>     Constant-power supplies will not cause this problem, but the
>     intensity of the brightest part of the arc will slowly deteriorate
>     with time.
>
>     Another factor is dirt. Either finger grease or perhaps some
>     accidental event such as a small, fried insect or a very small piece
>     of flaked-off black paint. In short, anything that absorbs light and
>     gets hot while attached in some way to the quartz envelope. Any "dark
>     spot" will cause local overheating and asymmetrical envelope failure.
>
>     A second possibility has to to with the connections between the bulb
>     and the housing. If the mating surfaces of these metal parts become
>     corroded, or oxidized, the contact resistance will increase as will
>     the working temperature. This also can lead to premature envelope
>     failure.
>
>     Even the alignment of the spherical rear mirror can be important. If
>     the reflected image of the arc is formed not to the side of the arc
>     itself but concentrated onto a small section of the envelope, this
>     can exacerbate any asymmetrical heating cause by dirt or
>     contamination.
>
>     The effect of these "operator" factors will vary somewhat depending
>     on the type of power supply used.
>
>     Therefore, it seems to me that the list of important variables should
>     include not just be the bulb manufacturer, the bulb type  and the
>     design of the power supply but the care with which the new bulb has
>     been installed and aligned.
>
>     And then there is the whole matter of variable-power Hg systems. We
>     usually use our Atto-arc 50W at about 20% power and the bulbs often
>     seem to last for over a thousand hours. More than enough brightness
>     to find  the part of your sample best suited to confocal analysis.
>
>     Finally, there is the matter of xenon arcs. These have all the same
>     problems mentioned above, but in addition, because they are reputed
>     to start off having a much higher gas pressure, any resulting
>     explosions are said to be much more impressive, in fact even
>     dangerous (albeit, without the additional nuisance of Hg
>     contamination). Has anyone had any experience here?
>
>     Cheers,
>
>     Jim Pawley
>
>
>      >Search the CONFOCAL archive at
>      >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>      >
>      >Hello everyone
>      >
>      >I think we are barking under the wrong tree.  Please correct me if i
>      >am wrong, but many explosions I read reported in this thread (and
>      >only one -in 25 years- which happened to me @50hrs) happened at way
>      >less than the rated hours.
>      >
>      >Shouldn't  we revise the myth that changing the lamp early protects
>      >from explosions?
>      >
>      >It seems to me that it is either due to material failure (in which
>      >case I think the lamp manufacturers should pay the damage bill) or
>      >too many ignitions, in which case the manufacturers should include
>      >this count in the specs.
>      >
>      >Regards,
>      >
>      >*********************************
>      >Stamatis Pagakis Ph.D.
>      >Biological Imaging Unit
>      >Biomedical Research Foundation, Academy of Athens,  Greece
>      >
>      >>On 4 May 2008, at 00:54, Russell Spear wrote:
>      >>Search the CONFOCAL archive at
>      >>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>      >>
>      >>So it isn't worth the risk to use them past rated hours.
>      >>
>      >>Russ
>      >>
>      >>
>      >>
>      >>Russell N. Spear
>      >>Sr. Research Specialist
>      >>Dept. of Plant Pathology
>      >>Univ. of Wisconsin
>      >>1630 Linden Dr.
>      >>Madison WI 53706
>      >>
>      >>voice 608.263.2093
>      >>fax     608.263.2626
>
>
>     --
>                    **********************************************
>     Prof. James B. Pawley,                                      Ph.
>     608-263-3147
>     Room 223, Zoology Research Building,
>     FAX  608-265-5315
>     1117 Johnson Ave., Madison, WI, 53706
>     [hidden email]
>     3D Microscopy of Living Cells Course, June 14-26, 2008, UBC,
>     Vancouver Canada
>     Info: http://www.3dcourse.ubc.ca/            Applications due by
>     March 15, 2008
>                    "If it ain't diffraction, it must be statistics." Anon.
>

--
Alison J. North, Ph.D.,
Research Assistant Professor and
Director of the Bio-Imaging Resource Center,
The Rockefeller University,
1230 York Avenue,
New York,
NY 10065.
Tel: office ++ 212 327 7488
Tel: lab     ++ 212 327 7486
Fax:         ++ 212 327 7489
Gary Laevsky-2 Gary Laevsky-2
Reply | Threaded
Open this post in threaded view
|

Re: HBO/Mercury lamp lifetimes ***COMMERCIAL RESPONSE***

In reply to this post by Alison J. North
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Alison and all,

CoolLed has a 3 channel version available in the states for $8,500 list.

Best,

 

Gary Laevsky, Ph.D.

Imaging Application Specialist

 

Andor Technology

discover new ways of seeing

 

[hidden email]

Cell          (774) 291 - 9992
Office       (860) 290 - 9211 x219
Fax          (860) 290 - 9566
Web:       www.andor.com

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Alison North
Sent: Wednesday, May 07, 2008 10:51 AM
To: [hidden email]
Subject: Re: HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Jim,

Oh you cruel man, you are mercilessly TAUNTING those of us based in the
USA who would love to buy your LEDs but are prevented from doing so by
exceptionally irritating patent restrictions (hello to the company
concerned - your LED system is prohibitively expensive!). Or am I
missing something and this situation has recently changed?  Do keep us
posted please?  And yes, I am intentionally sending this to the
listserver, so that other people based in this country don't get all
excited by your LEDs and have their hopes dashed, like us.  Plus, if I
am incorrect about the patent situation and we can in fact obtain an
LED-based lamp here at a reasonable price (without having to assemble
everything ourselves), we would like to know.

Thanks and best wishes,
Alison


James Beacher wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
> Commercial Interest
>  
>  
> Having followed this discussion, we can contain ourselves no
> longer.........*LEDs!*
>  
>  
>  
> JIM Beacher
> CoolLED Ltd
>  
>  
>
>     -----Original Message-----
>     *From:* "Confocal Microscopy List" <[hidden email]>
>     *Sent:* 07/05/2008 14:50
>     *To:* "[hidden email]"
<[hidden email]>
>     *Subject:* Re: HBO/Mercury lamp lifetimes
>
>
>     Search the CONFOCAL archive at
>     http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>     Hi all,
>
>     I would like to go back to an earlier post, and reiterate that
there
>     are several different types of power supplies. The
constant-current
>     ones increase the applied voltage as the "conductivity" of the
bulb
>     decreases. This means that the amount of power deposited in the
arc
>     slowly INCREASES as the electrodes age and eventually, this will
>     cause overheating, and softening of the fused-quartz envelope.
>     Constant-power supplies will not cause this problem, but the
>     intensity of the brightest part of the arc will slowly deteriorate
>     with time.
>
>     Another factor is dirt. Either finger grease or perhaps some
>     accidental event such as a small, fried insect or a very small
piece
>     of flaked-off black paint. In short, anything that absorbs light
and
>     gets hot while attached in some way to the quartz envelope. Any
"dark
>     spot" will cause local overheating and asymmetrical envelope
failure.
>
>     A second possibility has to to with the connections between the
bulb
>     and the housing. If the mating surfaces of these metal parts
become
>     corroded, or oxidized, the contact resistance will increase as
will
>     the working temperature. This also can lead to premature envelope
>     failure.
>
>     Even the alignment of the spherical rear mirror can be important.
If
>     the reflected image of the arc is formed not to the side of the
arc
>     itself but concentrated onto a small section of the envelope, this
>     can exacerbate any asymmetrical heating cause by dirt or
>     contamination.
>
>     The effect of these "operator" factors will vary somewhat
depending
>     on the type of power supply used.
>
>     Therefore, it seems to me that the list of important variables
should
>     include not just be the bulb manufacturer, the bulb type  and the
>     design of the power supply but the care with which the new bulb
has
>     been installed and aligned.
>
>     And then there is the whole matter of variable-power Hg systems.
We
>     usually use our Atto-arc 50W at about 20% power and the bulbs
often
>     seem to last for over a thousand hours. More than enough
brightness
>     to find  the part of your sample best suited to confocal analysis.
>
>     Finally, there is the matter of xenon arcs. These have all the
same
>     problems mentioned above, but in addition, because they are
reputed

>     to start off having a much higher gas pressure, any resulting
>     explosions are said to be much more impressive, in fact even
>     dangerous (albeit, without the additional nuisance of Hg
>     contamination). Has anyone had any experience here?
>
>     Cheers,
>
>     Jim Pawley
>
>
>      >Search the CONFOCAL archive at
>      >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>      >
>      >Hello everyone
>      >
>      >I think we are barking under the wrong tree.  Please correct me
if i
>      >am wrong, but many explosions I read reported in this thread
(and
>      >only one -in 25 years- which happened to me @50hrs) happened at
way
>      >less than the rated hours.
>      >
>      >Shouldn't  we revise the myth that changing the lamp early
protects
>      >from explosions?
>      >
>      >It seems to me that it is either due to material failure (in
which
>      >case I think the lamp manufacturers should pay the damage bill)
or
>      >too many ignitions, in which case the manufacturers should
include

>      >this count in the specs.
>      >
>      >Regards,
>      >
>      >*********************************
>      >Stamatis Pagakis Ph.D.
>      >Biological Imaging Unit
>      >Biomedical Research Foundation, Academy of Athens,  Greece
>      >
>      >>On 4 May 2008, at 00:54, Russell Spear wrote:
>      >>Search the CONFOCAL archive at
>      >>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>      >>
>      >>So it isn't worth the risk to use them past rated hours.
>      >>
>      >>Russ
>      >>
>      >>
>      >>
>      >>Russell N. Spear
>      >>Sr. Research Specialist
>      >>Dept. of Plant Pathology
>      >>Univ. of Wisconsin
>      >>1630 Linden Dr.
>      >>Madison WI 53706
>      >>
>      >>voice 608.263.2093
>      >>fax     608.263.2626
>
>
>     --
>                    **********************************************
>     Prof. James B. Pawley,                                      Ph.
>     608-263-3147
>     Room 223, Zoology Research Building,              
>     FAX  608-265-5315
>     1117 Johnson Ave., Madison, WI, 53706
>     [hidden email]
>     3D Microscopy of Living Cells Course, June 14-26, 2008, UBC,
>     Vancouver Canada
>     Info: http://www.3dcourse.ubc.ca/            Applications due by
>     March 15, 2008
>                    "If it ain't diffraction, it must be statistics."
Anon.
>

--
Alison J. North, Ph.D.,
Research Assistant Professor and
Director of the Bio-Imaging Resource Center,
The Rockefeller University,
1230 York Avenue,
New York,
NY 10065.
Tel: office ++ 212 327 7488
Tel: lab     ++ 212 327 7486
Fax:         ++ 212 327 7489
Martyn Reynolds Martyn Reynolds
Reply | Threaded
Open this post in threaded view
|

Re: HBO/Mercury lamp lifetimes

In reply to this post by James Pawley
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
* Commercial Interest *
 
Dear All,
Following recent posts here, I would like to contribute my experiences to the thread. First, the points made regarding the importance of the constant power driving of the short-arc lamps (both Hg and Xe) should not be underestimated. I cannot comment on failure rates of lamps installed in other manufacturers systems, but at Cairn we rarely encounter lamp explosions despite installation and operation of many hundreds of systems, many of which we know are heavily used. Needless to say, all our systems operate on a constant power basis (and the majority are Xe rather than Hg).
 
One cautionary point I would make is that users should stop operating a lamp immediately the output starts to fluctuate eratically. This does indeed indicate the lamp is reaching the very limits of its life, and the only explosions we have had reported are after this situation has been left unchecked.  Aside from the obvious health hazardof Hg contamination, I don't have any specificexperience of the relative hazard of Hg versus Xe lamp explosion, but in either case the damage caused by lamp failures is indeed quite spectacular- and usually costly to repair. Our housings completely enclose the lamp to maximise output stability, minimise light leakage, and protect against the explosion hazard. This means any explosion is completely contained, and while being impressively destructive internally, should not present a hazard to
the system operator.
 
In our hands the main problem with lamp aging is that the output intensity drops below an acceptable level rather than the envelope failing, and we know of systems being run many hundreds of hours beyond the rated life until this occurs. I would not advocate this as good practice though, and would advise exchanging lamps at their
rated lifetime.
 
My overall point is that whilst any pressurised system is potentially dangerous. With a few sensible precautions and a trained operator there is no significant risk in using these sources. This said, for an increasing number of applications, LED emitters are now of sufficient intensity that they make a good alternative for simple fluorescence applications (such as verification of transfection, or certain time-lapse imaging experiments). If you don't require the flexibility with wavelengths offered by a short-arc lamp, then LED systems such as our OptoLED or the PrecisExcite from CoolLED are increasingly useful alternatives.

Best Regards

Martyn
................................................
Dr A.M. Reynolds
www.cairn-research.co.uk
[hidden email]

Direct Line : + 44 (0)1795 594 503

 >>  Search the CONFOCAL archive at
 >>  http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

 >>  Hi all,

 >>  I would like to go back to an earlier post, and reiterate that there
 >>  are several different types of power supplies. The constant-current
 >>  ones increase the applied voltage as the "conductivity" of the bulb
 >>  decreases. This means that the amount of power deposited in the arc
 >>  slowly INCREASES as the electrodes age and eventually, this will
 >>  cause overheating, and softening of the fused-quartz envelope.
 >>  Constant-power supplies will not cause this problem, but the
 >>  intensity of the brightest part of the arc will slowly deteriorate
 >>  with time.

 >>  Another factor is dirt. Either finger grease or perhaps some
 >>  accidental event such as a small, fried insect or a very small piece
 >>  of flaked-off black paint. In short, anything that absorbs light and
 >>  gets hot while attached in some way to the quartz envelope. Any "dark
 >>  spot" will cause local overheating and asymmetrical envelope failure.

 >>  A second possibility has to to with the connections between the bulb
 >>  and the housing. If the mating surfaces of these metal parts become
 >>  corroded, or oxidized, the contact resistance will increase as will
 >>  the working temperature. This also can lead to premature envelope
 >>  failure.

 >>  Even the alignment of the spherical rear mirror can be important. If
 >>  the reflected image of the arc is formed not to the side of the arc
 >>  itself but concentrated onto a small section of the envelope, this
 >>  can exacerbate any asymmetrical heating cause by dirt or
 >>  contamination.

 >>  The effect of these "operator" factors will vary somewhat depending
 >>  on the type of power supply used.

 >>  Therefore, it seems to me that the list of important variables should
 >>  include not just be the bulb manufacturer, the bulb type  and the
 >>  design of the power supply but the care with which the new bulb has
 >>  been installed and aligned.

 >>  And then there is the whole matter of variable-power Hg systems. We
 >>  usually use our Atto-arc 50W at about 20% power and the bulbs often
 >>  seem to last for over a thousand hours. More than enough brightness
 >>  to find  the part of your sample best suited to confocal analysis.

 >>  Finally, there is the matter of xenon arcs. These have all the same
 >>  problems mentioned above, but in addition, because they are reputed
 >>  to start off having a much higher gas pressure, any resulting
 >>  explosions are said to be much more impressive, in fact even
 >>  dangerous (albeit, without the additional nuisance of Hg
 >>  contamination). Has anyone had any experience here?

 >>  Cheers,

 >>  Jim Pawley


 >>  >Search the CONFOCAL archive at
 >>  >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
 >>  >
 >>  >Hello everyone
 >>  >
 >>  >I think we are barking under the wrong tree.  Please correct me if i
 >>  >am wrong, but many explosions I read reported in this thread (and
 >>  >only one -in 25 years- which happened to me @50hrs) happened at way
 >>  >less than the rated hours.
 >>  >
 >>  >Shouldn't  we revise the myth that changing the lamp early protects
 >>  >from explosions?
 >>  >
 >>  >It seems to me that it is either due to material failure (in which
 >>  >case I think the lamp manufacturers should pay the damage bill) or
 >>  >too many ignitions, in which case the manufacturers should include
 >>  >this count in the specs.
 >>  >
 >>  >Regards,
 >>  >
 >>  >*********************************
 >>  >Stamatis Pagakis Ph.D.
 >>  >Biological Imaging Unit
 >>  >Biomedical Research Foundation, Academy of Athens,  Greece
 >>  >
 >>  >>On 4 May 2008, at 00:54, Russell Spear wrote:
 >>  >>Search the CONFOCAL archive at
 >>  >>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
 >>  >>
 >>  >>So it isn't worth the risk to use them past rated hours.
 >>  >>
 >>  >>Russ
 >>  >>
 >>  >>
 >>  >>
 >>  >>Russell N. Spear
 >>  >>Sr. Research Specialist
 >>  >>Dept. of Plant Pathology
 >>  >>Univ. of Wisconsin
 >>  >>1630 Linden Dr.
 >>  >>Madison WI 53706
 >>  >>
 >>  >>voice 608.263.2093
 >>  >>fax     608.263.2626


 >>  --
 >>                 **********************************************
 >>  Prof. James B. Pawley,                                   Ph.  608-263-3147
 >>  Room 223, Zoology Research Building,               
 >>  FAX  608-265-5315
 >>  1117 Johnson Ave., Madison, WI, 53706  
 >>  [hidden email]
 >>  3D Microscopy of Living Cells Course, June 14-26, 2008, UBC, Vancouver
 >>   Canada
 >>  Info: http://www.3dcourse.ubc.ca/         Applications due by March 15, 2008
 >>             "If it ain't diffraction, it must be statistics." Anon.
Gary Laevsky-2 Gary Laevsky-2
Reply | Threaded
Open this post in threaded view
|

Re: HBO/Mercury lamp lifetimes ***COMMERCIAL RESPONSE***

In reply to this post by Gary Laevsky-2
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

FYI, Andor Technology provides this in the states.

Best,

 

Gary Laevsky, Ph.D.

Imaging Application Specialist

 

Andor Technology

discover new ways of seeing

 

[hidden email]

Cell          (774) 291 - 9992
Office       (860) 290 - 9211 x219
Fax          (860) 290 - 9566
Web:       www.andor.com

-----Original Message-----
From: Gary Laevsky
Sent: Wednesday, May 07, 2008 11:15 AM
To: 'Confocal Microscopy List'
Subject: RE: HBO/Mercury lamp lifetimes ***COMMERCIAL RESPONSE***

Hi Alison and all,

CoolLed has a 3 channel version available in the states for $8,500 list.

Best,

 

Gary Laevsky, Ph.D.

Imaging Application Specialist

 

Andor Technology

discover new ways of seeing

 

[hidden email]

Cell          (774) 291 - 9992
Office       (860) 290 - 9211 x219
Fax          (860) 290 - 9566
Web:       www.andor.com

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Alison North
Sent: Wednesday, May 07, 2008 10:51 AM
To: [hidden email]
Subject: Re: HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Jim,

Oh you cruel man, you are mercilessly TAUNTING those of us based in the
USA who would love to buy your LEDs but are prevented from doing so by
exceptionally irritating patent restrictions (hello to the company
concerned - your LED system is prohibitively expensive!). Or am I
missing something and this situation has recently changed?  Do keep us
posted please?  And yes, I am intentionally sending this to the
listserver, so that other people based in this country don't get all
excited by your LEDs and have their hopes dashed, like us.  Plus, if I
am incorrect about the patent situation and we can in fact obtain an
LED-based lamp here at a reasonable price (without having to assemble
everything ourselves), we would like to know.

Thanks and best wishes,
Alison


James Beacher wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
> Commercial Interest
>  
>  
> Having followed this discussion, we can contain ourselves no
> longer.........*LEDs!*
>  
>  
>  
> JIM Beacher
> CoolLED Ltd
>  
>  
>
>     -----Original Message-----
>     *From:* "Confocal Microscopy List" <[hidden email]>
>     *Sent:* 07/05/2008 14:50
>     *To:* "[hidden email]"
<[hidden email]>
>     *Subject:* Re: HBO/Mercury lamp lifetimes
>
>
>     Search the CONFOCAL archive at
>     http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>     Hi all,
>
>     I would like to go back to an earlier post, and reiterate that
there
>     are several different types of power supplies. The
constant-current
>     ones increase the applied voltage as the "conductivity" of the
bulb
>     decreases. This means that the amount of power deposited in the
arc
>     slowly INCREASES as the electrodes age and eventually, this will
>     cause overheating, and softening of the fused-quartz envelope.
>     Constant-power supplies will not cause this problem, but the
>     intensity of the brightest part of the arc will slowly deteriorate
>     with time.
>
>     Another factor is dirt. Either finger grease or perhaps some
>     accidental event such as a small, fried insect or a very small
piece
>     of flaked-off black paint. In short, anything that absorbs light
and
>     gets hot while attached in some way to the quartz envelope. Any
"dark
>     spot" will cause local overheating and asymmetrical envelope
failure.
>
>     A second possibility has to to with the connections between the
bulb
>     and the housing. If the mating surfaces of these metal parts
become
>     corroded, or oxidized, the contact resistance will increase as
will
>     the working temperature. This also can lead to premature envelope
>     failure.
>
>     Even the alignment of the spherical rear mirror can be important.
If
>     the reflected image of the arc is formed not to the side of the
arc
>     itself but concentrated onto a small section of the envelope, this
>     can exacerbate any asymmetrical heating cause by dirt or
>     contamination.
>
>     The effect of these "operator" factors will vary somewhat
depending
>     on the type of power supply used.
>
>     Therefore, it seems to me that the list of important variables
should
>     include not just be the bulb manufacturer, the bulb type  and the
>     design of the power supply but the care with which the new bulb
has
>     been installed and aligned.
>
>     And then there is the whole matter of variable-power Hg systems.
We
>     usually use our Atto-arc 50W at about 20% power and the bulbs
often
>     seem to last for over a thousand hours. More than enough
brightness
>     to find  the part of your sample best suited to confocal analysis.
>
>     Finally, there is the matter of xenon arcs. These have all the
same
>     problems mentioned above, but in addition, because they are
reputed

>     to start off having a much higher gas pressure, any resulting
>     explosions are said to be much more impressive, in fact even
>     dangerous (albeit, without the additional nuisance of Hg
>     contamination). Has anyone had any experience here?
>
>     Cheers,
>
>     Jim Pawley
>
>
>      >Search the CONFOCAL archive at
>      >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>      >
>      >Hello everyone
>      >
>      >I think we are barking under the wrong tree.  Please correct me
if i
>      >am wrong, but many explosions I read reported in this thread
(and
>      >only one -in 25 years- which happened to me @50hrs) happened at
way
>      >less than the rated hours.
>      >
>      >Shouldn't  we revise the myth that changing the lamp early
protects
>      >from explosions?
>      >
>      >It seems to me that it is either due to material failure (in
which
>      >case I think the lamp manufacturers should pay the damage bill)
or
>      >too many ignitions, in which case the manufacturers should
include

>      >this count in the specs.
>      >
>      >Regards,
>      >
>      >*********************************
>      >Stamatis Pagakis Ph.D.
>      >Biological Imaging Unit
>      >Biomedical Research Foundation, Academy of Athens,  Greece
>      >
>      >>On 4 May 2008, at 00:54, Russell Spear wrote:
>      >>Search the CONFOCAL archive at
>      >>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>      >>
>      >>So it isn't worth the risk to use them past rated hours.
>      >>
>      >>Russ
>      >>
>      >>
>      >>
>      >>Russell N. Spear
>      >>Sr. Research Specialist
>      >>Dept. of Plant Pathology
>      >>Univ. of Wisconsin
>      >>1630 Linden Dr.
>      >>Madison WI 53706
>      >>
>      >>voice 608.263.2093
>      >>fax     608.263.2626
>
>
>     --
>                    **********************************************
>     Prof. James B. Pawley,                                      Ph.
>     608-263-3147
>     Room 223, Zoology Research Building,              
>     FAX  608-265-5315
>     1117 Johnson Ave., Madison, WI, 53706
>     [hidden email]
>     3D Microscopy of Living Cells Course, June 14-26, 2008, UBC,
>     Vancouver Canada
>     Info: http://www.3dcourse.ubc.ca/            Applications due by
>     March 15, 2008
>                    "If it ain't diffraction, it must be statistics."
Anon.
>

--
Alison J. North, Ph.D.,
Research Assistant Professor and
Director of the Bio-Imaging Resource Center,
The Rockefeller University,
1230 York Avenue,
New York,
NY 10065.
Tel: office ++ 212 327 7488
Tel: lab     ++ 212 327 7486
Fax:         ++ 212 327 7489
Alison J. North Alison J. North
Reply | Threaded
Open this post in threaded view
|

Re: HBO/Mercury lamp lifetimes ***COMMERCIAL RESPONSE***

In reply to this post by Gary Laevsky-2
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi all,

Is this true?  Thanks for the response Gary but we had an e-mail from
Jim Beacher recently informing us that CoolLED could not currently
supply it to the USA.  Jim, could we have clarification on this issue
please?  Is it that only particular models are available here?

Thanks,
Alison

P.S.  Since an Andor person has taken the trouble to help me out,
I feel I should be allowed to comment on an Andor product at this point
- we finally got round to demo-ing the iXon 512 EMCCD on our spinning
disk yesterday and now I see what all the fuss is about - fabulous
camera guys!!!





Gary Laevsky wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi Alison and all,
>
> CoolLed has a 3 channel version available in the states for $8,500 list.
>
> Best,
>
>  
>
> Gary Laevsky, Ph.D.
>
> Imaging Application Specialist
>
>  
>
> Andor Technology
>
> discover new ways of seeing
>
>  
>
> [hidden email]
>
> Cell          (774) 291 - 9992
> Office       (860) 290 - 9211 x219
> Fax          (860) 290 - 9566
> Web:       www.andor.com
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On
> Behalf Of Alison North
> Sent: Wednesday, May 07, 2008 10:51 AM
> To: [hidden email]
> Subject: Re: HBO/Mercury lamp lifetimes
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi Jim,
>
> Oh you cruel man, you are mercilessly TAUNTING those of us based in the
> USA who would love to buy your LEDs but are prevented from doing so by
> exceptionally irritating patent restrictions (hello to the company
> concerned - your LED system is prohibitively expensive!). Or am I
> missing something and this situation has recently changed?  Do keep us
> posted please?  And yes, I am intentionally sending this to the
> listserver, so that other people based in this country don't get all
> excited by your LEDs and have their hopes dashed, like us.  Plus, if I
> am incorrect about the patent situation and we can in fact obtain an
> LED-based lamp here at a reasonable price (without having to assemble
> everything ourselves), we would like to know.
>
> Thanks and best wishes,
> Alison
>
>
> James Beacher wrote:
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>> Commercial Interest
>>  
>>  
>> Having followed this discussion, we can contain ourselves no
>> longer.........*LEDs!*
>>  
>>  
>>  
>> JIM Beacher
>> CoolLED Ltd
>>  
>>  
>>
>>     -----Original Message-----
>>     *From:* "Confocal Microscopy List" <[hidden email]>
>>     *Sent:* 07/05/2008 14:50
>>     *To:* "[hidden email]"
> <[hidden email]>
>>     *Subject:* Re: HBO/Mercury lamp lifetimes
>>
>>
>>     Search the CONFOCAL archive at
>>     http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>>     Hi all,
>>
>>     I would like to go back to an earlier post, and reiterate that
> there
>>     are several different types of power supplies. The
> constant-current
>>     ones increase the applied voltage as the "conductivity" of the
> bulb
>>     decreases. This means that the amount of power deposited in the
> arc
>>     slowly INCREASES as the electrodes age and eventually, this will
>>     cause overheating, and softening of the fused-quartz envelope.
>>     Constant-power supplies will not cause this problem, but the
>>     intensity of the brightest part of the arc will slowly deteriorate
>>     with time.
>>
>>     Another factor is dirt. Either finger grease or perhaps some
>>     accidental event such as a small, fried insect or a very small
> piece
>>     of flaked-off black paint. In short, anything that absorbs light
> and
>>     gets hot while attached in some way to the quartz envelope. Any
> "dark
>>     spot" will cause local overheating and asymmetrical envelope
> failure.
>>     A second possibility has to to with the connections between the
> bulb
>>     and the housing. If the mating surfaces of these metal parts
> become
>>     corroded, or oxidized, the contact resistance will increase as
> will
>>     the working temperature. This also can lead to premature envelope
>>     failure.
>>
>>     Even the alignment of the spherical rear mirror can be important.
> If
>>     the reflected image of the arc is formed not to the side of the
> arc
>>     itself but concentrated onto a small section of the envelope, this
>>     can exacerbate any asymmetrical heating cause by dirt or
>>     contamination.
>>
>>     The effect of these "operator" factors will vary somewhat
> depending
>>     on the type of power supply used.
>>
>>     Therefore, it seems to me that the list of important variables
> should
>>     include not just be the bulb manufacturer, the bulb type  and the
>>     design of the power supply but the care with which the new bulb
> has
>>     been installed and aligned.
>>
>>     And then there is the whole matter of variable-power Hg systems.
> We
>>     usually use our Atto-arc 50W at about 20% power and the bulbs
> often
>>     seem to last for over a thousand hours. More than enough
> brightness
>>     to find  the part of your sample best suited to confocal analysis.
>>
>>     Finally, there is the matter of xenon arcs. These have all the
> same
>>     problems mentioned above, but in addition, because they are
> reputed
>>     to start off having a much higher gas pressure, any resulting
>>     explosions are said to be much more impressive, in fact even
>>     dangerous (albeit, without the additional nuisance of Hg
>>     contamination). Has anyone had any experience here?
>>
>>     Cheers,
>>
>>     Jim Pawley
>>
>>
>>      >Search the CONFOCAL archive at
>>      >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>      >
>>      >Hello everyone
>>      >
>>      >I think we are barking under the wrong tree.  Please correct me
> if i
>>      >am wrong, but many explosions I read reported in this thread
> (and
>>      >only one -in 25 years- which happened to me @50hrs) happened at
> way
>>      >less than the rated hours.
>>      >
>>      >Shouldn't  we revise the myth that changing the lamp early
> protects
>>      >from explosions?
>>      >
>>      >It seems to me that it is either due to material failure (in
> which
>>      >case I think the lamp manufacturers should pay the damage bill)
> or
>>      >too many ignitions, in which case the manufacturers should
> include
>>      >this count in the specs.
>>      >
>>      >Regards,
>>      >
>>      >*********************************
>>      >Stamatis Pagakis Ph.D.
>>      >Biological Imaging Unit
>>      >Biomedical Research Foundation, Academy of Athens,  Greece
>>      >
>>      >>On 4 May 2008, at 00:54, Russell Spear wrote:
>>      >>Search the CONFOCAL archive at
>>      >>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>      >>
>>      >>So it isn't worth the risk to use them past rated hours.
>>      >>
>>      >>Russ
>>      >>
>>      >>
>>      >>
>>      >>Russell N. Spear
>>      >>Sr. Research Specialist
>>      >>Dept. of Plant Pathology
>>      >>Univ. of Wisconsin
>>      >>1630 Linden Dr.
>>      >>Madison WI 53706
>>      >>
>>      >>voice 608.263.2093
>>      >>fax     608.263.2626
>>
>>
>>     --
>>                    **********************************************
>>     Prof. James B. Pawley,                                      Ph.
>>     608-263-3147
>>     Room 223, Zoology Research Building,              
>>     FAX  608-265-5315
>>     1117 Johnson Ave., Madison, WI, 53706
>>     [hidden email]
>>     3D Microscopy of Living Cells Course, June 14-26, 2008, UBC,
>>     Vancouver Canada
>>     Info: http://www.3dcourse.ubc.ca/            Applications due by
>>     March 15, 2008
>>                    "If it ain't diffraction, it must be statistics."
> Anon.
>

--
Alison J. North, Ph.D.,
Research Assistant Professor and
Director of the Bio-Imaging Resource Center,
The Rockefeller University,
1230 York Avenue,
New York,
NY 10065.
Tel: office ++ 212 327 7488
Tel: lab     ++ 212 327 7486
Fax:         ++ 212 327 7489
Gary Laevsky-2 Gary Laevsky-2
Reply | Threaded
Open this post in threaded view
|

Re: HBO/Mercury lamp lifetimes***COMMERCIAL RESPONSE*** CLARIFICATION

In reply to this post by Alison J. North
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hello List,

Checking internally, although we are planning to launch this product as
soon as possible, it is currently the topic of a licensing negotiation.

Jim Beacher of CoolLED will send a clarification.

I am sorry for the misinformation.

Best,

 

Gary Laevsky, Ph.D.

Imaging Application Specialist

 

Andor Technology

discover new ways of seeing

 

[hidden email]

Cell          (774) 291 - 9992
Office       (860) 290 - 9211 x219
Fax          (860) 290 - 9566
Web:       www.andor.com

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Alison North
Sent: Wednesday, May 07, 2008 10:51 AM
To: [hidden email]
Subject: Re: HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Jim,

Oh you cruel man, you are mercilessly TAUNTING those of us based in the
USA who would love to buy your LEDs but are prevented from doing so by
exceptionally irritating patent restrictions (hello to the company
concerned - your LED system is prohibitively expensive!). Or am I
missing something and this situation has recently changed?  Do keep us
posted please?  And yes, I am intentionally sending this to the
listserver, so that other people based in this country don't get all
excited by your LEDs and have their hopes dashed, like us.  Plus, if I
am incorrect about the patent situation and we can in fact obtain an
LED-based lamp here at a reasonable price (without having to assemble
everything ourselves), we would like to know.

Thanks and best wishes,
Alison


James Beacher wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
> Commercial Interest
>  
>  
> Having followed this discussion, we can contain ourselves no
> longer.........*LEDs!*
>  
>  
>  
> JIM Beacher
> CoolLED Ltd
>  
>  
>
>     -----Original Message-----
>     *From:* "Confocal Microscopy List" <[hidden email]>
>     *Sent:* 07/05/2008 14:50
>     *To:* "[hidden email]"
<[hidden email]>
>     *Subject:* Re: HBO/Mercury lamp lifetimes
>
>
>     Search the CONFOCAL archive at
>     http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>     Hi all,
>
>     I would like to go back to an earlier post, and reiterate that
there
>     are several different types of power supplies. The
constant-current
>     ones increase the applied voltage as the "conductivity" of the
bulb
>     decreases. This means that the amount of power deposited in the
arc
>     slowly INCREASES as the electrodes age and eventually, this will
>     cause overheating, and softening of the fused-quartz envelope.
>     Constant-power supplies will not cause this problem, but the
>     intensity of the brightest part of the arc will slowly deteriorate
>     with time.
>
>     Another factor is dirt. Either finger grease or perhaps some
>     accidental event such as a small, fried insect or a very small
piece
>     of flaked-off black paint. In short, anything that absorbs light
and
>     gets hot while attached in some way to the quartz envelope. Any
"dark
>     spot" will cause local overheating and asymmetrical envelope
failure.
>
>     A second possibility has to to with the connections between the
bulb
>     and the housing. If the mating surfaces of these metal parts
become
>     corroded, or oxidized, the contact resistance will increase as
will
>     the working temperature. This also can lead to premature envelope
>     failure.
>
>     Even the alignment of the spherical rear mirror can be important.
If
>     the reflected image of the arc is formed not to the side of the
arc
>     itself but concentrated onto a small section of the envelope, this
>     can exacerbate any asymmetrical heating cause by dirt or
>     contamination.
>
>     The effect of these "operator" factors will vary somewhat
depending
>     on the type of power supply used.
>
>     Therefore, it seems to me that the list of important variables
should
>     include not just be the bulb manufacturer, the bulb type  and the
>     design of the power supply but the care with which the new bulb
has
>     been installed and aligned.
>
>     And then there is the whole matter of variable-power Hg systems.
We
>     usually use our Atto-arc 50W at about 20% power and the bulbs
often
>     seem to last for over a thousand hours. More than enough
brightness
>     to find  the part of your sample best suited to confocal analysis.
>
>     Finally, there is the matter of xenon arcs. These have all the
same
>     problems mentioned above, but in addition, because they are
reputed

>     to start off having a much higher gas pressure, any resulting
>     explosions are said to be much more impressive, in fact even
>     dangerous (albeit, without the additional nuisance of Hg
>     contamination). Has anyone had any experience here?
>
>     Cheers,
>
>     Jim Pawley
>
>
>      >Search the CONFOCAL archive at
>      >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>      >
>      >Hello everyone
>      >
>      >I think we are barking under the wrong tree.  Please correct me
if i
>      >am wrong, but many explosions I read reported in this thread
(and
>      >only one -in 25 years- which happened to me @50hrs) happened at
way
>      >less than the rated hours.
>      >
>      >Shouldn't  we revise the myth that changing the lamp early
protects
>      >from explosions?
>      >
>      >It seems to me that it is either due to material failure (in
which
>      >case I think the lamp manufacturers should pay the damage bill)
or
>      >too many ignitions, in which case the manufacturers should
include

>      >this count in the specs.
>      >
>      >Regards,
>      >
>      >*********************************
>      >Stamatis Pagakis Ph.D.
>      >Biological Imaging Unit
>      >Biomedical Research Foundation, Academy of Athens,  Greece
>      >
>      >>On 4 May 2008, at 00:54, Russell Spear wrote:
>      >>Search the CONFOCAL archive at
>      >>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>      >>
>      >>So it isn't worth the risk to use them past rated hours.
>      >>
>      >>Russ
>      >>
>      >>
>      >>
>      >>Russell N. Spear
>      >>Sr. Research Specialist
>      >>Dept. of Plant Pathology
>      >>Univ. of Wisconsin
>      >>1630 Linden Dr.
>      >>Madison WI 53706
>      >>
>      >>voice 608.263.2093
>      >>fax     608.263.2626
>
>
>     --
>                    **********************************************
>     Prof. James B. Pawley,                                      Ph.
>     608-263-3147
>     Room 223, Zoology Research Building,              
>     FAX  608-265-5315
>     1117 Johnson Ave., Madison, WI, 53706
>     [hidden email]
>     3D Microscopy of Living Cells Course, June 14-26, 2008, UBC,
>     Vancouver Canada
>     Info: http://www.3dcourse.ubc.ca/            Applications due by
>     March 15, 2008
>                    "If it ain't diffraction, it must be statistics."
Anon.
>

--
Alison J. North, Ph.D.,
Research Assistant Professor and
Director of the Bio-Imaging Resource Center,
The Rockefeller University,
1230 York Avenue,
New York,
NY 10065.
Tel: office ++ 212 327 7488
Tel: lab     ++ 212 327 7486
Fax:         ++ 212 327 7489
Michael Cammer Michael Cammer
Reply | Threaded
Open this post in threaded view
|

Re: HBO/Mercury lamp lifetimes ***COMMERCIAL RESPONSE***

In reply to this post by Alison J. North
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Just out of curiosity, is there anyone on this list who has witnessed
an Hg lamp used for microscopy exploding?  Or is this a subject for
an article at http://snopes.com/ ?
____________________________________________________________________________
Michael Cammer, Senior Light Microscopist, Analytical Imaging
Facility, Albert Einstein College of Medicine, 1300 Morris Park
Avenue, Bronx, NY  10461
URLs: microscopy http://www.aecom.yu.edu/aif/ and art http://coxcammer.com/
Ignatius, Mike Ignatius, Mike
Reply | Threaded
Open this post in threaded view
|

Re: HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

No urban myth, this.  Sounds like a pistol shot or firecracker.  I was
three inches away, quite a fright.  

I concur with Dr. Pawley's comments - check the bulb from time to time
for deposition and don't focus the light through the light if you have
that type of collector.

If the bulb doesn't fire readily, check it! If you smell electrically
type smoke, turn it off!

Mike Ignatius

Molecular Probes/Invitrogen

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Michael Cammer
Sent: Wednesday, May 07, 2008 10:44 AM
To: [hidden email]
Subject: Re: HBO/Mercury lamp lifetimes ***COMMERCIAL RESPONSE***

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Just out of curiosity, is there anyone on this list who has witnessed
an Hg lamp used for microscopy exploding?  Or is this a subject for
an article at http://snopes.com/ ?
________________________________________________________________________
____
Michael Cammer, Senior Light Microscopist, Analytical Imaging
Facility, Albert Einstein College of Medicine, 1300 Morris Park
Avenue, Bronx, NY  10461
URLs: microscopy http://www.aecom.yu.edu/aif/ and art
http://coxcammer.com/
Renato A. Mortara Renato A. Mortara
Reply | Threaded
Open this post in threaded view
|

RES: HBO/Mercury lamp lifetimes ***COMMERCIAL RESPONSE***

In reply to this post by Michael Cammer
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I had a Hg lamp exploded on me. Fortunately I wear glasses that prevented
major damage.

Renato Mortara
Dr. Renato A. Mortara
Division of Parasitology
Departament of Microbiology, Immunology and Parasitology
UNIFESP - Escola Paulista de Medicina
Rua Botucatu, 862 6o andar
04023-062 São Paulo SP Brazil
Phone: 55 11 5579-8306
FAX:  55 11 5571-1095
email: [hidden email]
home page: www.ecb.epm.br/~renato

-----Mensagem original-----
De: Confocal Microscopy List [mailto:[hidden email]] Em nome
de Michael Cammer
Enviada em: quarta-feira, 7 de maio de 2008 14:44
Para: [hidden email]
Assunto: Re: HBO/Mercury lamp lifetimes ***COMMERCIAL RESPONSE***

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Just out of curiosity, is there anyone on this list who has witnessed an Hg
lamp used for microscopy exploding?  Or is this a subject for an article at
http://snopes.com/ ?
____________________________________________________________________________
Michael Cammer, Senior Light Microscopist, Analytical Imaging Facility,
Albert Einstein College of Medicine, 1300 Morris Park Avenue, Bronx, NY
10461
URLs: microscopy http://www.aecom.yu.edu/aif/ and art http://coxcammer.com/
123