http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Thanks for all your helpful suggestions. I thought the following might be interesting: I used my laser power meter to check the power density for different lenses and light sources that we have. At 740 nm (Chroma SP-106 excitation filter is 740±35) No lens - 43 mW with Hg and 58 mW with Halogen (3200K) 20x/0.45 Achroplan Ph2 - 79%T 40x/1.2 W C Apochromat - 81%T 63x/1.4 oil PlanApo - 67%T At 774 nm (peak absorption for IRDye 800) No lens - 40 mW with Hg and 50 mW with Halogen (3200K) 20x/0.45 Achroplan Ph2 - 50%T 40x/1.2 W C Apochromat - 60%T 63x/1.4 oil PlanApo - 42%T There is a lot more light out there with the Hg than you might think even though the Halogen is better. The %Ts at 740 nm are pretty similar to the "typical" curves for these lenses found on the Zeiss website. The drop off when going up to 774 nm was quite a bit more than expected. Licor tech support does not have information on the power density of the laser in the Odyssey but they thought it was pretty low. We put the power meter in the Odyssey and never recorded more than 3 mW during the scan. So compared to the Odyssey we have plenty of light. Kate |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Kate, what type of scanner is this Odyssey? If it's a point-scanning device, then 3 mW per point is a lot (nice for FRET or FRAP). So you would have to consider the amount of scanned pixels in your comparison, maybe 1024 times 1024... times 3 mW. Michael > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Thanks for all your helpful suggestions. I thought the following might be > interesting: > > I used my laser power meter to check the power density for different > lenses > and light sources that we have. > > At 740 nm (Chroma SP-106 excitation filter is 740±35) > > No lens - 43 mW with Hg and 58 mW with Halogen (3200K) > 20x/0.45 Achroplan Ph2 - 79%T > 40x/1.2 W C Apochromat - 81%T > 63x/1.4 oil PlanApo - 67%T > > At 774 nm (peak absorption for IRDye 800) > > No lens - 40 mW with Hg and 50 mW with Halogen (3200K) > 20x/0.45 Achroplan Ph2 - 50%T > 40x/1.2 W C Apochromat - 60%T > 63x/1.4 oil PlanApo - 42%T > > There is a lot more light out there with the Hg than you might think even > though the Halogen is better. The %Ts at 740 nm are pretty similar to the > "typical" curves for these lenses found on the Zeiss website. The drop off > when going up to 774 nm was quite a bit more than expected. > > Licor tech support does not have information on the power density of the > laser in the Odyssey but they thought it was pretty low. We put the power > meter in the Odyssey and never recorded more than 3 mW during the scan. > > So compared to the Odyssey we have plenty of light. > > Kate |
|
|
In reply to this post by Kate Luby-Phelps
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal You are absolutely right. I realized that after I had posted the previous message. Probably it is point scanning although not diffraction limited. Meanwhile, the FOV of the lens would be much larger. So the power density is potentially much greater in the scanner. Thanks. Kate |
| Free forum by Nabble | Edit this page |