Aleksander Szczurek |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi All, I was wondering if anybody of you knows where to purchase immersion oils of different refractive index (RI). We are currently in need of immersion oils with RI in the following range: 1.510, 1.512, 1.514, 1.516. I would be very grateful for any hints and experience from your side. greetings Aleksander Szczurek |
![]() ![]() |
Boehm, Ulrike (NIH/NCI) [F] |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Aleksander, I used to order immersion oil from Cargille (http://www.cargille.com/immerliq.shtml ). Best regards, Ulrike ____________________ Ulrike Boehm, Ph.D. Laboratory of Receptor Biology & Gene Expression NCI, National Institutes of Health 41 Medlars Drive Building 41, Room B303 Bethesda, MD-20892, United States p (office): +1 (240) 760 6581 p (lab): +1 (240) 760 6686 f: +1 (240) 541 4450 http://ulrikeboehm.org/ [hidden email] -----Original Message----- From: Aleksander Szczurek [mailto:[hidden email]] Sent: Friday, September 22, 2017 5:17 PM To: [hidden email] Subject: Immersion oil with various refractive index ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi All, I was wondering if anybody of you knows where to purchase immersion oils of different refractive index (RI). We are currently in need of immersion oils with RI in the following range: 1.510, 1.512, 1.514, 1.516. I would be very grateful for any hints and experience from your side. greetings Aleksander Sz czurek |
![]() ![]() |
Eric Marino |
![]() |
In reply to this post by Aleksander Szczurek
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** You can try … http://www.cargille.com/immeroil.shtml?gclid=CjwKEAjw6ZLOBRCLyNXAwfCPmlQSJADHPIJbzGBNklUsGHXOfVbns2QHJ1BcYsF5e6N8qenm0gaHARoC-ODw_wcB Eric Marino [hidden email] On 9/22/17, 5:16 PM, "Confocal Microscopy List on behalf of Aleksander Szczurek" <[hidden email] on behalf of [hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi All, I was wondering if anybody of you knows where to purchase immersion oils of different refractive index (RI). We are currently in need of immersion oils with RI in the following range: 1.510, 1.512, 1.514, 1.516. I would be very grateful for any hints and experience from your side. greetings Aleksander Szczurek |
![]() ![]() |
Craig Brideau |
![]() |
In reply to this post by Aleksander Szczurek
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Cargille sells a range of various index oils. According to an old Cargille brochure I happen to have, 'Series A' oils go from 1.460 to 1.640 in steps of 0.002! A full set used to run you ~$430 USD back in 2012, and half sets were about half the price. Craig On Fri, Sep 22, 2017 at 3:16 PM, Aleksander Szczurek < [hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi All, > I was wondering if anybody of you knows where to purchase immersion oils of > different refractive index (RI). We are currently in need of immersion oils > with RI in the following range: 1.510, 1.512, 1.514, 1.516. I would be very > grateful for any hints and experience from your side. > > greetings > Aleksander Szczurek > |
![]() ![]() |
David Elliott |
![]() |
In reply to this post by Aleksander Szczurek
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** As a cheaper option you could also try thiodiethanol mixed with water, as described in the following article: https://biochimie.umontreal.ca/wp-content/uploads/sites/37/2012/08/Thiodiethanol-new-water-mounting-media-2007.pdf |
![]() ![]() |
Benjamin Smith |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Wow, that thiodiethanol paper is awesome, thanks for the heads-up! Although, I can't imaging it smells too great (and the SDS for it seems to agree). On Fri, Sep 22, 2017 at 3:29 PM, David Elliott <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > As a cheaper option you could also try thiodiethanol mixed with water, as > described in the following article: > > https://biochimie.umontreal.ca/wp-content/uploads/sites/ > 37/2012/08/Thiodiethanol-new-water-mounting-media-2007.pdf > -- Benjamin E. Smith, Ph. D. Imaging Specialist, Vision Science University of California, Berkeley 195 Life Sciences Addition Berkeley, CA 94720-3200 Tel (510) 642-9712 Fax (510) 643-6791 e-mail: [hidden email] http://vision.berkeley.edu/?page_id=5635 <http://vision.berkeley.edu/> |
![]() ![]() |
Zdenek Svindrych-2 |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Benjamin, don't worry, it's not a thiol, it does not smell at all. But it's a useful stuff for sure. About the price of glycerol, but less viscous and higher RI. Not hygroscopic. I use it a lot for index matching, but haven't actually used it as a substitute for immersion oil. You can dilute it with water to the required refractive index, but the water will evaporate over time and the RI will increase. I would mix it with something with lower water pressure, such as propylene glycol (but maybe something less hygroscopic and less viscous)... Best, zdenek -- Zdenek Svindrych, Ph.D. W.M. Keck Center for Cellular Imaging (PLSB 003) Department of Biology,University of Virginia 409 McCormick Rd, Charlottesville, VA-22904 http://www.kcci.virginia.edu/ tel: 434-982-4869 ---------- Původní e-mail ---------- Od: Benjamin E Smith <[hidden email]> Komu: [hidden email] Datum: 22. 9. 2017 20:30:12 Předmět: Re: Immersion oil with various refractive index "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Wow, that thiodiethanol paper is awesome, thanks for the heads-up! Although, I can't imaging it smells too great (and the SDS for it seems to agree). On Fri, Sep 22, 2017 at 3:29 PM, David Elliott <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > As a cheaper option you could also try thiodiethanol mixed with water, as > described in the following article: > > https://biochimie.umontreal.ca/wp-content/uploads/sites/ > 37/2012/08/Thiodiethanol-new-water-mounting-media-2007.pdf > -- Benjamin E. Smith, Ph. D. Imaging Specialist, Vision Science University of California, Berkeley 195 Life Sciences Addition Berkeley, CA 94720-3200 Tel (510) 642-9712 Fax (510) 643-6791 e-mail: [hidden email] http://vision.berkeley.edu/?page_id=5635 <http://vision.berkeley.edu/> " |
![]() ![]() |
Tim Feinstein |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Agreed that TDE is a great way to mix up a non-hardening mount with whatever RI you need. It has no smell. Ethanedithiol, on the other hand, smells like the end of the world. It's like beta mercaptoethanol's meaner big brother. They don't warn you about that stuff in the papers that describe tetracysteine motif labeling. Best, T Timothy Feinstein, Ph.D. Research Scientist University of Pittsburgh Department of Developmental Biology -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email] Sent: Friday, September 22, 2017 8:48 PM To: [hidden email] Subject: Re: Immersion oil with various refractive index ***** To join, leave or search the confocal microscopy listserv, go to: https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=01%7C01%7Ctnf8%40PITT.EDU%7Cb2aab68a7b9c4d11b05608d5021ccc09%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1&sdata=syyYmxPfHErGFqyWzw%2B9zheAT3XHJGsdshBLX%2BvwTJQ%3D&reserved=0 Post images on https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=01%7C01%7Ctnf8%40PITT.EDU%7Cb2aab68a7b9c4d11b05608d5021ccc09%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1&sdata=R4ExCA0hu0QBvjAmJD%2FxMby3GmKmqgPIXKU%2FhpX%2BHaI%3D&reserved=0 and include the link in your posting. ***** Benjamin, don't worry, it's not a thiol, it does not smell at all. But it's a useful stuff for sure. About the price of glycerol, but less viscous and higher RI. Not hygroscopic. I use it a lot for index matching, but haven't actually used it as a substitute for immersion oil. You can dilute it with water to the required refractive index, but the water will evaporate over time and the RI will increase. I would mix it with something with lower water pressure, such as propylene glycol (but maybe something less hygroscopic and less viscous)... Best, zdenek -- Zdenek Svindrych, Ph.D. W.M. Keck Center for Cellular Imaging (PLSB 003) Department of Biology,University of Virginia 409 McCormick Rd, Charlottesville, VA-22904 https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.kcci.virginia.edu%2F&data=01%7C01%7Ctnf8%40PITT.EDU%7Cb2aab68a7b9c4d11b05608d5021ccc09%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1&sdata=eT%2Fh1ThyJXMeOUh8KEoUD4CBxkHG%2FpDq1tBxoWFUTUE%3D&reserved=0 tel: 434-982-4869 ---------- Původní e-mail ---------- Od: Benjamin E Smith <[hidden email]> Komu: [hidden email] Datum: 22. 9. 2017 20:30:12 Předmět: Re: Immersion oil with various refractive index "***** To join, leave or search the confocal microscopy listserv, go to: https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=01%7C01%7Ctnf8%40PITT.EDU%7Cb2aab68a7b9c4d11b05608d5021ccc09%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1&sdata=syyYmxPfHErGFqyWzw%2B9zheAT3XHJGsdshBLX%2BvwTJQ%3D&reserved=0 Post images on https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=01%7C01%7Ctnf8%40PITT.EDU%7Cb2aab68a7b9c4d11b05608d5021ccc09%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1&sdata=R4ExCA0hu0QBvjAmJD%2FxMby3GmKmqgPIXKU%2FhpX%2BHaI%3D&reserved=0 and include the link in your posting. ***** Wow, that thiodiethanol paper is awesome, thanks for the heads-up! Although, I can't imaging it smells too great (and the SDS for it seems to agree). On Fri, Sep 22, 2017 at 3:29 PM, David Elliott <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists. > umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=01%7C01%7Ctnf8%4 > 0PITT.EDU%7Cb2aab68a7b9c4d11b05608d5021ccc09%7C9ef9f489e0a04eeb87cc3a5 > 26112fd0d%7C1&sdata=syyYmxPfHErGFqyWzw%2B9zheAT3XHJGsdshBLX%2BvwTJQ%3D > &reserved=0 Post images on > https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=01%7C01%7Ctnf8%40PITT.EDU%7Cb2aab68a7b9c4d11b05608d5021ccc09%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1&sdata=R4ExCA0hu0QBvjAmJD%2FxMby3GmKmqgPIXKU%2FhpX%2BHaI%3D&reserved=0 and include the link in your posting. > ***** > > As a cheaper option you could also try thiodiethanol mixed with water, > as described in the following article: > > https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fbioch > imie.umontreal.ca%2Fwp-content%2Fuploads%2Fsites%2F&data=01%7C01%7Ctnf > 8%40PITT.EDU%7Cb2aab68a7b9c4d11b05608d5021ccc09%7C9ef9f489e0a04eeb87cc > 3a526112fd0d%7C1&sdata=tDT4wuI2UrFlMB0lIWExIaJXiW7mOwYxzfS2WKMHvC0%3D& > reserved=0 37/2012/08/Thiodiethanol-new-water-mounting-media-2007.pdf > -- Benjamin E. Smith, Ph. D. Imaging Specialist, Vision Science University of California, Berkeley 195 Life Sciences Addition Berkeley, CA 94720-3200 Tel (510) 642-9712 Fax (510) 643-6791 e-mail: [hidden email] https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fvision.berkeley.edu%2F%3Fpage_id%3D5635&data=01%7C01%7Ctnf8%40PITT.EDU%7Cb2aab68a7b9c4d11b05608d5021ccc09%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1&sdata=WrZIm9h9NMRDDFwrW54y3WdcMyRFjEtT71m1LaMRUdA%3D&reserved=0 <https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fvision.berkeley.edu%2F&data=01%7C01%7Ctnf8%40PITT.EDU%7Cb2aab68a7b9c4d11b05608d5021ccc09%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1&sdata=DwOBOscJ32Ym4Xi2MtBrJrgydb1v%2BLX58ZnpuTc2hz4%3D&reserved=0> " |
![]() ![]() |
Barbara Foster |
![]() |
In reply to this post by Aleksander Szczurek
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi, Alexander, Try The McCrone Group (www.mccrone.com). They have carried RI standards for many years and are a highly trusted source. Also, remember that RI is very wavelength and temperature dependent. To be accurate to the third decimal place, as you have indicated in your message, use a narrowband filter over the light source and, preferrably, a temperature controlled stage. Most RI's are given nD (sodium D line) and 25o C. Hope this was helpful. (Caveat: No commercial interest) Barbara Foster, President & Chief Consultant Microscopy/Microscopy Education ... "Education, not JustTraining" 7101 Royal Glen Trail, Suite A - McKinney, TX 75070 P: 972-924-5310 W: www.MicroscopyEducation.com Microscopy/Microscopy Education is a division of The Microscopy & Imaging Place, Inc. At 09:57 AM 9/22/2017, you wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >Post images on http://www.imgur.com and include the link in your posting. >***** > >Hi All, >I was wondering if anybody of you knows where to purchase immersion oils of >different refractive index (RI). We are currently in need of immersion oils >with RI in the following range: 1.510, 1.512, 1.514, 1.516. I would be very >grateful for any hints and experience from your side. > >greetings >Aleksander Szczurek |
![]() ![]() |
Aleksander Szczurek |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear All, thank you very much for a valuable input. I knew this is going to be a good place to ask :-) Wish you all the best in your experiments! Aleks 2017-09-27 22:11 GMT+02:00 Barbara Foster <[hidden email]>: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi, Alexander, > > Try The McCrone Group (www.mccrone.com). They have carried RI standards > for many years and are a highly trusted source. > > Also, remember that RI is very wavelength and temperature dependent. To > be accurate to the third decimal place, as you have indicated in your > message, use a narrowband filter over the light source and, preferrably, a > temperature controlled stage. Most RI's are given nD (sodium D line) and > 25o C. > > Hope this was helpful. (Caveat: No commercial interest) > Barbara Foster, President & Chief Consultant > Microscopy/Microscopy Education ... "Education, not JustTraining" > 7101 Royal Glen Trail, Suite A - McKinney, TX 75070 > P: 972-924-5310 W: www.MicroscopyEducation.com > Microscopy/Microscopy Education is a division of The Microscopy & Imaging > Place, Inc. > At 09:57 AM 9/22/2017, you wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Hi All, >> I was wondering if anybody of you knows where to purchase immersion oils >> of >> different refractive index (RI). We are currently in need of immersion >> oils >> with RI in the following range: 1.510, 1.512, 1.514, 1.516. I would be >> very >> grateful for any hints and experience from your side. >> >> greetings >> Aleksander Szczurek >> > |